Hi!

Can anyone give me a rough value of the thermistor resistance that is
brought out to the front panel of a Fluke 732A DC Reference Standard (lower
left-hand side)?  The thermistor is for the user to monitor the temperature  of
the oven that contains the voltage reference.  I see a resistance that
starts out at 6600 Ohms (unit was nearly cold) and climbs and levels out at
about 8400 Ohms after a day, or so.  Is this reasonable?  The  thermistor is
given in the parts list as RT2, a Fenwal JA41J1 (no longer made)  and on the
schematic as a 10K @ 25C unit.  So this doesn't make  sense.

Does anyone have any experience with this?  Or is just simpler to  replace
the thermistor with a 100 Ohm RTD and be done with it?  Or would  this drive
the calibration labs nuts?

Any references or advice would be appreciated.

Tom Leedy
Clarksburg, MD

Hi! Can anyone give me a rough value of the thermistor resistance that is brought out to the front panel of a Fluke 732A DC Reference Standard (lower left-hand side)? The thermistor is for the user to monitor the temperature of the oven that contains the voltage reference. I see a resistance that starts out at 6600 Ohms (unit was nearly cold) and climbs and levels out at about 8400 Ohms after a day, or so. Is this reasonable? The thermistor is given in the parts list as RT2, a Fenwal JA41J1 (no longer made) and on the schematic as a 10K @ 25C unit. So this doesn't make sense. Does anyone have any experience with this? Or is just simpler to replace the thermistor with a 100 Ohm RTD and be done with it? Or would this drive the calibration labs nuts? Any references or advice would be appreciated. Tom Leedy Clarksburg, MD
RK
Rob Klein
Thu, Jun 28, 2012 11:45 PM

Hi Tom,

Definitely not good! The thermistor is supposed to be an NTC, so seeing
the value
rise from cold means there's something very wrong indeed.

If you want to replace it, this might be a good alterative:
http://www.digikey.nl/product-detail/en/192-103LET-A01/480-3158-ND/1886044

Kind regards,
Rob Klein.

Op 29-6-2012 1:04, Leedyt@aol.com schreef:

Hi!

Can anyone give me a rough value of the thermistor resistance that is
brought out to the front panel of a Fluke 732A DC Reference Standard (lower
left-hand side)?  The thermistor is for the user to monitor the temperature  of
the oven that contains the voltage reference.  I see a resistance that
starts out at 6600 Ohms (unit was nearly cold) and climbs and levels out at
about 8400 Ohms after a day, or so.  Is this reasonable?  The  thermistor is
given in the parts list as RT2, a Fenwal JA41J1 (no longer made)  and on the
schematic as a 10K @ 25C unit.  So this doesn't make  sense.

Does anyone have any experience with this?  Or is just simpler to  replace
the thermistor with a 100 Ohm RTD and be done with it?  Or would  this drive
the calibration labs nuts?

Any references or advice would be appreciated.

Tom Leedy
Clarksburg, MD


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi Tom, Definitely not good! The thermistor is supposed to be an NTC, so seeing the value rise from cold means there's something very wrong indeed. If you want to replace it, this might be a good alterative: http://www.digikey.nl/product-detail/en/192-103LET-A01/480-3158-ND/1886044 Kind regards, Rob Klein. Op 29-6-2012 1:04, Leedyt@aol.com schreef: > > Hi! > > Can anyone give me a rough value of the thermistor resistance that is > brought out to the front panel of a Fluke 732A DC Reference Standard (lower > left-hand side)? The thermistor is for the user to monitor the temperature of > the oven that contains the voltage reference. I see a resistance that > starts out at 6600 Ohms (unit was nearly cold) and climbs and levels out at > about 8400 Ohms after a day, or so. Is this reasonable? The thermistor is > given in the parts list as RT2, a Fenwal JA41J1 (no longer made) and on the > schematic as a 10K @ 25C unit. So this doesn't make sense. > > Does anyone have any experience with this? Or is just simpler to replace > the thermistor with a 100 Ohm RTD and be done with it? Or would this drive > the calibration labs nuts? > > Any references or advice would be appreciated. > > Tom Leedy > Clarksburg, MD > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >
MV
Mitch Van Ochten
Fri, Jun 29, 2012 1:42 AM

Hi Tom,

According to the Fluke 732A manual page 2-6 the nominal value is between
3k - 4k with the oven at normal operating temperature.  Mine checks 3.545 k
and has been within a few ohms of that reading each year for five years. You
may have a heater problem.  There is a protective thermal fuse which can go
open, in series with the heating elements.  It's purpose is to prevent a
runaway oven from cooking everything.  Once it opens you will never get any
heating until it is replaced.

Regards,

mitch

----- Original Message -----
From: Leedyt@aol.com
To: volt-nuts@febo.com
Sent: Thursday, June 28, 2012 7:04 PM
Subject: [volt-nuts] Thermistor value for 732A Reference 10V Standard

Hi!

Can anyone give me a rough value of the thermistor resistance that is
brought out to the front panel of a Fluke 732A DC Reference Standard
(lower
left-hand side)?  The thermistor is for the user to monitor the
temperature  of
the oven that contains the voltage reference.  I see a resistance that
starts out at 6600 Ohms (unit was nearly cold) and climbs and levels out
at
about 8400 Ohms after a day, or so.  Is this reasonable?  The  thermistor
is
given in the parts list as RT2, a Fenwal JA41J1 (no longer made)  and on
the
schematic as a 10K @ 25C unit.  So this doesn't make  sense.

Does anyone have any experience with this?  Or is just simpler to  replace
the thermistor with a 100 Ohm RTD and be done with it?  Or would  this
drive
the calibration labs nuts?

Any references or advice would be appreciated.

Tom Leedy
Clarksburg, MD


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi Tom, According to the Fluke 732A manual page 2-6 the nominal value is between 3k - 4k with the oven at normal operating temperature. Mine checks 3.545 k and has been within a few ohms of that reading each year for five years. You may have a heater problem. There is a protective thermal fuse which can go open, in series with the heating elements. It's purpose is to prevent a runaway oven from cooking everything. Once it opens you will never get any heating until it is replaced. Regards, mitch ----- Original Message ----- From: <Leedyt@aol.com> To: <volt-nuts@febo.com> Sent: Thursday, June 28, 2012 7:04 PM Subject: [volt-nuts] Thermistor value for 732A Reference 10V Standard > > > Hi! > > Can anyone give me a rough value of the thermistor resistance that is > brought out to the front panel of a Fluke 732A DC Reference Standard > (lower > left-hand side)? The thermistor is for the user to monitor the > temperature of > the oven that contains the voltage reference. I see a resistance that > starts out at 6600 Ohms (unit was nearly cold) and climbs and levels out > at > about 8400 Ohms after a day, or so. Is this reasonable? The thermistor > is > given in the parts list as RT2, a Fenwal JA41J1 (no longer made) and on > the > schematic as a 10K @ 25C unit. So this doesn't make sense. > > Does anyone have any experience with this? Or is just simpler to replace > the thermistor with a 100 Ohm RTD and be done with it? Or would this > drive > the calibration labs nuts? > > Any references or advice would be appreciated. > > Tom Leedy > Clarksburg, MD > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
D
DaveH
Fri, Jun 29, 2012 4:02 AM

Are you reading this in-circuit and operating? The current flowing through
the resistor will skew the reading.

Stupid question I know but had to be asked...

DaveH

-----Original Message-----
From: volt-nuts-bounces@febo.com
[mailto:volt-nuts-bounces@febo.com] On Behalf Of Mitch Van Ochten
Sent: Thursday, June 28, 2012 18:43
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Thermistor value for 732A Reference
10V Standard

Hi Tom,

According to the Fluke 732A manual page 2-6 the nominal value
is between
3k - 4k with the oven at normal operating temperature.  Mine
checks 3.545 k
and has been within a few ohms of that reading each year for
five years. You
may have a heater problem.  There is a protective thermal
fuse which can go
open, in series with the heating elements.  It's purpose is
to prevent a
runaway oven from cooking everything.  Once it opens you will
never get any
heating until it is replaced.

Regards,

mitch

----- Original Message -----
From: Leedyt@aol.com
To: volt-nuts@febo.com
Sent: Thursday, June 28, 2012 7:04 PM
Subject: [volt-nuts] Thermistor value for 732A Reference 10V Standard

Hi!

Can anyone give me a rough value of the thermistor

resistance that is

brought out to the front panel of a Fluke 732A DC Reference

Standard

(lower
left-hand side)?  The thermistor is for the user to monitor the
temperature  of
the oven that contains the voltage reference.  I see a

resistance that

starts out at 6600 Ohms (unit was nearly cold) and climbs

and levels out

at
about 8400 Ohms after a day, or so.  Is this reasonable?

The  thermistor

is
given in the parts list as RT2, a Fenwal JA41J1 (no longer

made)  and on

the
schematic as a 10K @ 25C unit.  So this doesn't make  sense.

Does anyone have any experience with this?  Or is just

simpler to  replace

the thermistor with a 100 Ohm RTD and be done with it?  Or

would  this

drive
the calibration labs nuts?

Any references or advice would be appreciated.

Tom Leedy
Clarksburg, MD


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Are you reading this in-circuit and operating? The current flowing through the resistor will skew the reading. Stupid question I know but had to be asked... DaveH > -----Original Message----- > From: volt-nuts-bounces@febo.com > [mailto:volt-nuts-bounces@febo.com] On Behalf Of Mitch Van Ochten > Sent: Thursday, June 28, 2012 18:43 > To: Discussion of precise voltage measurement > Subject: Re: [volt-nuts] Thermistor value for 732A Reference > 10V Standard > > Hi Tom, > > According to the Fluke 732A manual page 2-6 the nominal value > is between > 3k - 4k with the oven at normal operating temperature. Mine > checks 3.545 k > and has been within a few ohms of that reading each year for > five years. You > may have a heater problem. There is a protective thermal > fuse which can go > open, in series with the heating elements. It's purpose is > to prevent a > runaway oven from cooking everything. Once it opens you will > never get any > heating until it is replaced. > > > Regards, > > mitch > > > > ----- Original Message ----- > From: <Leedyt@aol.com> > To: <volt-nuts@febo.com> > Sent: Thursday, June 28, 2012 7:04 PM > Subject: [volt-nuts] Thermistor value for 732A Reference 10V Standard > > > > > > > > Hi! > > > > Can anyone give me a rough value of the thermistor > resistance that is > > brought out to the front panel of a Fluke 732A DC Reference > Standard > > (lower > > left-hand side)? The thermistor is for the user to monitor the > > temperature of > > the oven that contains the voltage reference. I see a > resistance that > > starts out at 6600 Ohms (unit was nearly cold) and climbs > and levels out > > at > > about 8400 Ohms after a day, or so. Is this reasonable? > The thermistor > > is > > given in the parts list as RT2, a Fenwal JA41J1 (no longer > made) and on > > the > > schematic as a 10K @ 25C unit. So this doesn't make sense. > > > > Does anyone have any experience with this? Or is just > simpler to replace > > the thermistor with a 100 Ohm RTD and be done with it? Or > would this > > drive > > the calibration labs nuts? > > > > Any references or advice would be appreciated. > > > > Tom Leedy > > Clarksburg, MD > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@febo.com > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > and follow the instructions there. > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
TL
Tom Leedy
Fri, Jun 29, 2012 4:34 AM

Rob Klein <rob.klein@...> writes:

Hi Tom,

Definitely not good! The thermistor is supposed to be an NTC, so seeing
the value
rise from cold means there's something very wrong indeed.

If you want to replace it, this might be a good alterative:
http://www.digikey.nl/product-detail/en/192-103LET-A01/480-3158-ND/1886044

Kind regards,
Rob Klein.

Op 29-6-2012 1:04, Leedyt@... schreef:

Hi!

Can anyone give me a rough value of the thermistor resistance that is
brought out to the front panel of a Fluke 732A DC Reference Standard (lower
left-hand side)?  The thermistor is for the user to monitor the temperature

of

the oven that contains the voltage reference.  I see a resistance that
starts out at 6600 Ohms (unit was nearly cold) and climbs and levels out at
about 8400 Ohms after a day, or so.  Is this reasonable?  The  thermistor is
given in the parts list as RT2, a Fenwal JA41J1 (no longer made)  and on the
schematic as a 10K @ 25C unit.  So this doesn't make  sense.

Does anyone have any experience with this?  Or is just simpler to  replace
the thermistor with a 100 Ohm RTD and be done with it?  Or would  this drive
the calibration labs nuts?

Any references or advice would be appreciated.

Tom Leedy
Clarksburg, MD


volt-nuts mailing list -- volt-nuts@...
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-

nuts

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@...
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Many thanks to all for the advice.  I suspect that somebody "replaced" the
original thermistor with another unit and didn't know what they were doing.
Since the unit seems to put out 10 Vdc (+/- less than a ppm or so) at ambient
temperature, I suspect that the internal oven is operating correctly.  But I
might investigate further by raising and lowering its ambient a few degrees and
checking to see that temperature of the oven remains constant.

Has anyone actually measured the internal temperature of the oven (in degrees
C)?  Since the operating temperature of the 732A is spec'd at 40 C, I suspect
that it is about 45-50 C, or so.  And I also suspect that temperature stability
of the oven is more important than absolute accuracy.

I'll let you know what I find.  Thanks to all again -- I didn't expect the
large outpouring of knowledge!

Regards -- Tom

Rob Klein <rob.klein@...> writes: > > Hi Tom, > > Definitely not good! The thermistor is supposed to be an NTC, so seeing > the value > rise from cold means there's something very wrong indeed. > > If you want to replace it, this might be a good alterative: > http://www.digikey.nl/product-detail/en/192-103LET-A01/480-3158-ND/1886044 > > Kind regards, > Rob Klein. > > Op 29-6-2012 1:04, Leedyt@... schreef: > > > > Hi! > > > > Can anyone give me a rough value of the thermistor resistance that is > > brought out to the front panel of a Fluke 732A DC Reference Standard (lower > > left-hand side)? The thermistor is for the user to monitor the temperature of > > the oven that contains the voltage reference. I see a resistance that > > starts out at 6600 Ohms (unit was nearly cold) and climbs and levels out at > > about 8400 Ohms after a day, or so. Is this reasonable? The thermistor is > > given in the parts list as RT2, a Fenwal JA41J1 (no longer made) and on the > > schematic as a 10K @ 25C unit. So this doesn't make sense. > > > > Does anyone have any experience with this? Or is just simpler to replace > > the thermistor with a 100 Ohm RTD and be done with it? Or would this drive > > the calibration labs nuts? > > > > Any references or advice would be appreciated. > > > > Tom Leedy > > Clarksburg, MD > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@... > > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt- nuts > > and follow the instructions there. > > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@... > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > Many thanks to all for the advice. I suspect that somebody "replaced" the original thermistor with another unit and didn't know what they were doing. Since the unit seems to put out 10 Vdc (+/- less than a ppm or so) at ambient temperature, I suspect that the internal oven is operating correctly. But I might investigate further by raising and lowering its ambient a few degrees and checking to see that temperature of the oven remains constant. Has anyone actually measured the internal temperature of the oven (in degrees C)? Since the operating temperature of the 732A is spec'd at 40 C, I suspect that it is about 45-50 C, or so. And I also suspect that temperature stability of the oven is more important than absolute accuracy. I'll let you know what I find. Thanks to all again -- I didn't expect the large outpouring of knowledge! Regards -- Tom
RK
Rob Klein
Fri, Jun 29, 2012 10:29 AM

Op 29-6-2012 6:34, Tom Leedy schreef:

Has anyone actually measured the internal temperature of the oven (in degrees
C)?  Since the operating temperature of the 732A is spec'd at 40 C, I suspect
that it is about 45-50 C, or so.

According to the manual it's 48 ± 2°C

And I also suspect that temperature stability
of the oven is more important than absolute accuracy.

You'd be correct :-)

I didn't expect the large outpouring of knowledge!

Nobody expects the large outpouring of knowledge! (with apologies to
M. Python)

Rob.

Op 29-6-2012 6:34, Tom Leedy schreef: > Has anyone actually measured the internal temperature of the oven (in degrees > C)? Since the operating temperature of the 732A is spec'd at 40 C, I suspect > that it is about 45-50 C, or so. According to the manual it's 48 ± 2°C > And I also suspect that temperature stability > of the oven is more important than absolute accuracy. You'd be correct :-) > > I didn't expect the large outpouring of knowledge! *Nobody* expects the large outpouring of knowledge! (with apologies to M. Python) Rob.
MV
Mitch Van Ochten
Fri, Jun 29, 2012 11:03 AM

I think monitoring the 1V or 1.1V is a better test over temperature than
monitoring the 10V.  It's quite stable even without the heater.

----- Original Message -----
From: "Rob Klein" rob.klein@smalldesign.nl
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, June 29, 2012 6:29 AM
Subject: Re: [volt-nuts] Thermistor value for 732A Reference 10V Standard

Op 29-6-2012 6:34, Tom Leedy schreef:

Has anyone actually measured the internal temperature of the oven (in
degrees
C)?  Since the operating temperature of the 732A is spec'd at 40 C, I
suspect
that it is about 45-50 C, or so.

According to the manual it's 48 ± 2°C

And I also suspect that temperature stability
of the oven is more important than absolute accuracy.

You'd be correct :-)

I didn't expect the large outpouring of knowledge!

Nobody expects the large outpouring of knowledge! (with apologies to
M. Python)

Rob.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

I think monitoring the 1V or 1.1V is a better test over temperature than monitoring the 10V. It's quite stable even without the heater. ----- Original Message ----- From: "Rob Klein" <rob.klein@smalldesign.nl> To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> Sent: Friday, June 29, 2012 6:29 AM Subject: Re: [volt-nuts] Thermistor value for 732A Reference 10V Standard Op 29-6-2012 6:34, Tom Leedy schreef: > Has anyone actually measured the internal temperature of the oven (in > degrees > C)? Since the operating temperature of the 732A is spec'd at 40 C, I > suspect > that it is about 45-50 C, or so. According to the manual it's 48 ± 2°C > And I also suspect that temperature stability > of the oven is more important than absolute accuracy. You'd be correct :-) > > I didn't expect the large outpouring of knowledge! *Nobody* expects the large outpouring of knowledge! (with apologies to M. Python) Rob. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
W
WB6BNQ
Fri, Jun 29, 2012 11:33 AM

Mitch,

Quite the contrary.  The 10 volt output is the master output and therefore the
most stable.  The others are derived from dividers and thus add additional error
terms due to the tempco of the divider string and any possible loading presented
to those outputs.

Bill....WB6BNQ

Mitch Van Ochten wrote:

I think monitoring the 1V or 1.1V is a better test over temperature than
monitoring the 10V.  It's quite stable even without the heater.

----- Original Message -----
From: "Rob Klein" rob.klein@smalldesign.nl
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, June 29, 2012 6:29 AM
Subject: Re: [volt-nuts] Thermistor value for 732A Reference 10V Standard

Op 29-6-2012 6:34, Tom Leedy schreef:

Has anyone actually measured the internal temperature of the oven (in
degrees
C)?  Since the operating temperature of the 732A is spec'd at 40 C, I
suspect
that it is about 45-50 C, or so.

According to the manual it's 48 ± 2°C

And I also suspect that temperature stability
of the oven is more important than absolute accuracy.

You'd be correct :-)

I didn't expect the large outpouring of knowledge!

Nobody expects the large outpouring of knowledge! (with apologies to
M. Python)

Rob.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Mitch, Quite the contrary. The 10 volt output is the master output and therefore the most stable. The others are derived from dividers and thus add additional error terms due to the tempco of the divider string and any possible loading presented to those outputs. Bill....WB6BNQ Mitch Van Ochten wrote: > I think monitoring the 1V or 1.1V is a better test over temperature than > monitoring the 10V. It's quite stable even without the heater. > > ----- Original Message ----- > From: "Rob Klein" <rob.klein@smalldesign.nl> > To: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > Sent: Friday, June 29, 2012 6:29 AM > Subject: Re: [volt-nuts] Thermistor value for 732A Reference 10V Standard > > Op 29-6-2012 6:34, Tom Leedy schreef: > > Has anyone actually measured the internal temperature of the oven (in > > degrees > > C)? Since the operating temperature of the 732A is spec'd at 40 C, I > > suspect > > that it is about 45-50 C, or so. > > According to the manual it's 48 ± 2°C > > > And I also suspect that temperature stability > > of the oven is more important than absolute accuracy. > > You'd be correct :-) > > > > I didn't expect the large outpouring of knowledge! > > *Nobody* expects the large outpouring of knowledge! (with apologies to > M. Python) > > Rob. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
RK
Rob Klein
Fri, Jun 29, 2012 12:02 PM

Bill,

I had to read it twice myself, but that was Mitch' point. When you want
to check the performance
of the oven, you want to measure something that has a decent bit of
variation over temperature.

Rob.

Op 29-6-2012 13:33, WB6BNQ schreef:

Mitch,

Quite the contrary.  The 10 volt output is the master output and therefore the
most stable.  The others are derived from dividers and thus add additional error
terms due to the tempco of the divider string and any possible loading presented
to those outputs.

Bill....WB6BNQ

Mitch Van Ochten wrote:

I think monitoring the 1V or 1.1V is a better test over temperature than
monitoring the 10V.  It's quite stable even without the heater.

Bill, I had to read it twice myself, but that was Mitch' point. When you want to check the performance of the oven, you want to measure something that has a decent bit of variation over temperature. Rob. Op 29-6-2012 13:33, WB6BNQ schreef: > Mitch, > > Quite the contrary. The 10 volt output is the master output and therefore the > most stable. The others are derived from dividers and thus add additional error > terms due to the tempco of the divider string and any possible loading presented > to those outputs. > > Bill....WB6BNQ > > Mitch Van Ochten wrote: > >> I think monitoring the 1V or 1.1V is a better test over temperature than >> monitoring the 10V. It's quite stable even without the heater. >> >>
W
WB6BNQ
Fri, Jun 29, 2012 7:53 PM

Hi Rob,

Yes, now I see the point !  My apologies to Mitch for misunderstanding his point.

Bill....WB6BNQ

Rob Klein wrote:

Bill,

I had to read it twice myself, but that was Mitch' point. When you want
to check the performance
of the oven, you want to measure something that has a decent bit of
variation over temperature.

Rob.

Op 29-6-2012 13:33, WB6BNQ schreef:

Mitch,

Quite the contrary.  The 10 volt output is the master output and therefore the
most stable.  The others are derived from dividers and thus add additional error
terms due to the tempco of the divider string and any possible loading presented
to those outputs.

Bill....WB6BNQ

Mitch Van Ochten wrote:

I think monitoring the 1V or 1.1V is a better test over temperature than
monitoring the 10V.  It's quite stable even without the heater.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi Rob, Yes, now I see the point ! My apologies to Mitch for misunderstanding his point. Bill....WB6BNQ Rob Klein wrote: > Bill, > > I had to read it twice myself, but that was Mitch' point. When you want > to check the performance > of the oven, you want to measure something that has a decent bit of > variation over temperature. > > Rob. > > Op 29-6-2012 13:33, WB6BNQ schreef: > > Mitch, > > > > Quite the contrary. The 10 volt output is the master output and therefore the > > most stable. The others are derived from dividers and thus add additional error > > terms due to the tempco of the divider string and any possible loading presented > > to those outputs. > > > > Bill....WB6BNQ > > > > Mitch Van Ochten wrote: > > > >> I think monitoring the 1V or 1.1V is a better test over temperature than > >> monitoring the 10V. It's quite stable even without the heater. > >> > >> > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.