Hi!
Can anyone give me a rough value of the thermistor resistance that is
brought out to the front panel of a Fluke 732A DC Reference Standard (lower
left-hand side)? The thermistor is for the user to monitor the temperature of
the oven that contains the voltage reference. I see a resistance that
starts out at 6600 Ohms (unit was nearly cold) and climbs and levels out at
about 8400 Ohms after a day, or so. Is this reasonable? The thermistor is
given in the parts list as RT2, a Fenwal JA41J1 (no longer made) and on the
schematic as a 10K @ 25C unit. So this doesn't make sense.
Does anyone have any experience with this? Or is just simpler to replace
the thermistor with a 100 Ohm RTD and be done with it? Or would this drive
the calibration labs nuts?
Any references or advice would be appreciated.
Tom Leedy
Clarksburg, MD
Hi Tom,
Definitely not good! The thermistor is supposed to be an NTC, so seeing
the value
rise from cold means there's something very wrong indeed.
If you want to replace it, this might be a good alterative:
http://www.digikey.nl/product-detail/en/192-103LET-A01/480-3158-ND/1886044
Kind regards,
Rob Klein.
Op 29-6-2012 1:04, Leedyt@aol.com schreef:
Hi!
Can anyone give me a rough value of the thermistor resistance that is
brought out to the front panel of a Fluke 732A DC Reference Standard (lower
left-hand side)? The thermistor is for the user to monitor the temperature of
the oven that contains the voltage reference. I see a resistance that
starts out at 6600 Ohms (unit was nearly cold) and climbs and levels out at
about 8400 Ohms after a day, or so. Is this reasonable? The thermistor is
given in the parts list as RT2, a Fenwal JA41J1 (no longer made) and on the
schematic as a 10K @ 25C unit. So this doesn't make sense.
Does anyone have any experience with this? Or is just simpler to replace
the thermistor with a 100 Ohm RTD and be done with it? Or would this drive
the calibration labs nuts?
Any references or advice would be appreciated.
Tom Leedy
Clarksburg, MD
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Hi Tom,
According to the Fluke 732A manual page 2-6 the nominal value is between
3k - 4k with the oven at normal operating temperature. Mine checks 3.545 k
and has been within a few ohms of that reading each year for five years. You
may have a heater problem. There is a protective thermal fuse which can go
open, in series with the heating elements. It's purpose is to prevent a
runaway oven from cooking everything. Once it opens you will never get any
heating until it is replaced.
Regards,
mitch
----- Original Message -----
From: Leedyt@aol.com
To: volt-nuts@febo.com
Sent: Thursday, June 28, 2012 7:04 PM
Subject: [volt-nuts] Thermistor value for 732A Reference 10V Standard
Hi!
Can anyone give me a rough value of the thermistor resistance that is
brought out to the front panel of a Fluke 732A DC Reference Standard
(lower
left-hand side)? The thermistor is for the user to monitor the
temperature of
the oven that contains the voltage reference. I see a resistance that
starts out at 6600 Ohms (unit was nearly cold) and climbs and levels out
at
about 8400 Ohms after a day, or so. Is this reasonable? The thermistor
is
given in the parts list as RT2, a Fenwal JA41J1 (no longer made) and on
the
schematic as a 10K @ 25C unit. So this doesn't make sense.
Does anyone have any experience with this? Or is just simpler to replace
the thermistor with a 100 Ohm RTD and be done with it? Or would this
drive
the calibration labs nuts?
Any references or advice would be appreciated.
Tom Leedy
Clarksburg, MD
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Are you reading this in-circuit and operating? The current flowing through
the resistor will skew the reading.
Stupid question I know but had to be asked...
DaveH
-----Original Message-----
From: volt-nuts-bounces@febo.com
[mailto:volt-nuts-bounces@febo.com] On Behalf Of Mitch Van Ochten
Sent: Thursday, June 28, 2012 18:43
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Thermistor value for 732A Reference
10V Standard
Hi Tom,
According to the Fluke 732A manual page 2-6 the nominal value
is between
3k - 4k with the oven at normal operating temperature. Mine
checks 3.545 k
and has been within a few ohms of that reading each year for
five years. You
may have a heater problem. There is a protective thermal
fuse which can go
open, in series with the heating elements. It's purpose is
to prevent a
runaway oven from cooking everything. Once it opens you will
never get any
heating until it is replaced.
Regards,
mitch
----- Original Message -----
From: Leedyt@aol.com
To: volt-nuts@febo.com
Sent: Thursday, June 28, 2012 7:04 PM
Subject: [volt-nuts] Thermistor value for 732A Reference 10V Standard
Hi!
Can anyone give me a rough value of the thermistor
resistance that is
brought out to the front panel of a Fluke 732A DC Reference
Standard
(lower
left-hand side)? The thermistor is for the user to monitor the
temperature of
the oven that contains the voltage reference. I see a
resistance that
starts out at 6600 Ohms (unit was nearly cold) and climbs
and levels out
at
about 8400 Ohms after a day, or so. Is this reasonable?
The thermistor
is
given in the parts list as RT2, a Fenwal JA41J1 (no longer
made) and on
the
schematic as a 10K @ 25C unit. So this doesn't make sense.
Does anyone have any experience with this? Or is just
simpler to replace
the thermistor with a 100 Ohm RTD and be done with it? Or
would this
drive
the calibration labs nuts?
Any references or advice would be appreciated.
Tom Leedy
Clarksburg, MD
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Rob Klein <rob.klein@...> writes:
Hi Tom,
Definitely not good! The thermistor is supposed to be an NTC, so seeing
the value
rise from cold means there's something very wrong indeed.
If you want to replace it, this might be a good alterative:
http://www.digikey.nl/product-detail/en/192-103LET-A01/480-3158-ND/1886044
Kind regards,
Rob Klein.
Op 29-6-2012 1:04, Leedyt@... schreef:
Hi!
Can anyone give me a rough value of the thermistor resistance that is
brought out to the front panel of a Fluke 732A DC Reference Standard (lower
left-hand side)? The thermistor is for the user to monitor the temperature
of
the oven that contains the voltage reference. I see a resistance that
starts out at 6600 Ohms (unit was nearly cold) and climbs and levels out at
about 8400 Ohms after a day, or so. Is this reasonable? The thermistor is
given in the parts list as RT2, a Fenwal JA41J1 (no longer made) and on the
schematic as a 10K @ 25C unit. So this doesn't make sense.
Does anyone have any experience with this? Or is just simpler to replace
the thermistor with a 100 Ohm RTD and be done with it? Or would this drive
the calibration labs nuts?
Any references or advice would be appreciated.
Tom Leedy
Clarksburg, MD
volt-nuts mailing list -- volt-nuts@...
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-
nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@...
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Many thanks to all for the advice. I suspect that somebody "replaced" the
original thermistor with another unit and didn't know what they were doing.
Since the unit seems to put out 10 Vdc (+/- less than a ppm or so) at ambient
temperature, I suspect that the internal oven is operating correctly. But I
might investigate further by raising and lowering its ambient a few degrees and
checking to see that temperature of the oven remains constant.
Has anyone actually measured the internal temperature of the oven (in degrees
C)? Since the operating temperature of the 732A is spec'd at 40 C, I suspect
that it is about 45-50 C, or so. And I also suspect that temperature stability
of the oven is more important than absolute accuracy.
I'll let you know what I find. Thanks to all again -- I didn't expect the
large outpouring of knowledge!
Regards -- Tom
Op 29-6-2012 6:34, Tom Leedy schreef:
Has anyone actually measured the internal temperature of the oven (in degrees
C)? Since the operating temperature of the 732A is spec'd at 40 C, I suspect
that it is about 45-50 C, or so.
According to the manual it's 48 ± 2°C
And I also suspect that temperature stability
of the oven is more important than absolute accuracy.
You'd be correct :-)
I didn't expect the large outpouring of knowledge!
Nobody expects the large outpouring of knowledge! (with apologies to
M. Python)
Rob.
I think monitoring the 1V or 1.1V is a better test over temperature than
monitoring the 10V. It's quite stable even without the heater.
----- Original Message -----
From: "Rob Klein" rob.klein@smalldesign.nl
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, June 29, 2012 6:29 AM
Subject: Re: [volt-nuts] Thermistor value for 732A Reference 10V Standard
Op 29-6-2012 6:34, Tom Leedy schreef:
Has anyone actually measured the internal temperature of the oven (in
degrees
C)? Since the operating temperature of the 732A is spec'd at 40 C, I
suspect
that it is about 45-50 C, or so.
According to the manual it's 48 ± 2°C
And I also suspect that temperature stability
of the oven is more important than absolute accuracy.
You'd be correct :-)
I didn't expect the large outpouring of knowledge!
Nobody expects the large outpouring of knowledge! (with apologies to
M. Python)
Rob.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Mitch,
Quite the contrary. The 10 volt output is the master output and therefore the
most stable. The others are derived from dividers and thus add additional error
terms due to the tempco of the divider string and any possible loading presented
to those outputs.
Bill....WB6BNQ
Mitch Van Ochten wrote:
I think monitoring the 1V or 1.1V is a better test over temperature than
monitoring the 10V. It's quite stable even without the heater.
----- Original Message -----
From: "Rob Klein" rob.klein@smalldesign.nl
To: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Friday, June 29, 2012 6:29 AM
Subject: Re: [volt-nuts] Thermistor value for 732A Reference 10V Standard
Op 29-6-2012 6:34, Tom Leedy schreef:
Has anyone actually measured the internal temperature of the oven (in
degrees
C)? Since the operating temperature of the 732A is spec'd at 40 C, I
suspect
that it is about 45-50 C, or so.
According to the manual it's 48 ± 2°C
And I also suspect that temperature stability
of the oven is more important than absolute accuracy.
You'd be correct :-)
I didn't expect the large outpouring of knowledge!
Nobody expects the large outpouring of knowledge! (with apologies to
M. Python)
Rob.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Bill,
I had to read it twice myself, but that was Mitch' point. When you want
to check the performance
of the oven, you want to measure something that has a decent bit of
variation over temperature.
Rob.
Op 29-6-2012 13:33, WB6BNQ schreef:
Mitch,
Quite the contrary. The 10 volt output is the master output and therefore the
most stable. The others are derived from dividers and thus add additional error
terms due to the tempco of the divider string and any possible loading presented
to those outputs.
Bill....WB6BNQ
Mitch Van Ochten wrote:
I think monitoring the 1V or 1.1V is a better test over temperature than
monitoring the 10V. It's quite stable even without the heater.
Hi Rob,
Yes, now I see the point ! My apologies to Mitch for misunderstanding his point.
Bill....WB6BNQ
Rob Klein wrote:
Bill,
I had to read it twice myself, but that was Mitch' point. When you want
to check the performance
of the oven, you want to measure something that has a decent bit of
variation over temperature.
Rob.
Op 29-6-2012 13:33, WB6BNQ schreef:
Mitch,
Quite the contrary. The 10 volt output is the master output and therefore the
most stable. The others are derived from dividers and thus add additional error
terms due to the tempco of the divider string and any possible loading presented
to those outputs.
Bill....WB6BNQ
Mitch Van Ochten wrote:
I think monitoring the 1V or 1.1V is a better test over temperature than
monitoring the 10V. It's quite stable even without the heater.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.