Hi,

Some years ago I was working with a conductance measuring product that maintained sample cells to 2 millidegrees, and 4 millidegrees front to back of the temperature bath for a stability held week by week at any dialed setting, but accuracy between dialed temp and actual was only expected to be about a degree. I would expect something similar for this, maybe not the 2 millidegree stability though, that was hard to achieve.

Date: Fri, 29 Jun 2012 12:29:19 +0200
From: rob.klein@smalldesign.nl
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Thermistor value for 732A Reference 10V Standard

Op 29-6-2012 6:34, Tom Leedy schreef:

Has anyone actually measured the internal temperature of the oven (in degrees
C)?  Since the operating temperature of the 732A is spec'd at 40 C, I suspect
that it is about 45-50 C, or so.

According to the manual it's 48 ± 2°C

And I also suspect that temperature stability
of the oven is more important than absolute accuracy.

You'd be correct :-)

I didn't expect the large outpouring of knowledge!

Nobody expects the large outpouring of knowledge! (with apologies to
M. Python)

Rob.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi, Some years ago I was working with a conductance measuring product that maintained sample cells to 2 millidegrees, and 4 millidegrees front to back of the temperature bath for a stability held week by week at any dialed setting, but accuracy between dialed temp and actual was only expected to be about a degree. I would expect something similar for this, maybe not the 2 millidegree stability though, that was hard to achieve. > Date: Fri, 29 Jun 2012 12:29:19 +0200 > From: rob.klein@smalldesign.nl > To: volt-nuts@febo.com > Subject: Re: [volt-nuts] Thermistor value for 732A Reference 10V Standard > > Op 29-6-2012 6:34, Tom Leedy schreef: > > Has anyone actually measured the internal temperature of the oven (in degrees > > C)? Since the operating temperature of the 732A is spec'd at 40 C, I suspect > > that it is about 45-50 C, or so. > > According to the manual it's 48 ± 2°C > > > And I also suspect that temperature stability > > of the oven is more important than absolute accuracy. > > You'd be correct :-) > > > > I didn't expect the large outpouring of knowledge! > > *Nobody* expects the large outpouring of knowledge! (with apologies to > M. Python) > > > Rob. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.