RS
Richard Solomon
Sat, Sep 9, 2017 9:24 PM
I have a working Z3801A, it has had an AC Supply added to it and the RS422 port has been reconfigured to RS232.
I downloaded and installed Z38XX, configured both the Port and Device to the recommended parameters:
Data Rate: 19,200
Parity: Odd
Data Bits: 7
All I can get, looking at the Device Window is:
Send *IDN ?
Received
Detected Device Undefined
I am assuming the Z801A and the Computer are not talking to each other, but I am at a loss as to what to do next.
I tried a different (known working) RS232 to USB adapter and get the same results.
Any ides what I can do next ?
Thanks, Dick, W1KSZ
Sent from Outlookhttp://aka.ms/weboutlook
I have a working Z3801A, it has had an AC Supply added to it and the RS422 port has been reconfigured to RS232.
I downloaded and installed Z38XX, configured both the Port and Device to the recommended parameters:
Data Rate: 19,200
Parity: Odd
Data Bits: 7
All I can get, looking at the Device Window is:
Send *IDN ?
Received
Detected Device Undefined
I am assuming the Z801A and the Computer are not talking to each other, but I am at a loss as to what to do next.
I tried a different (known working) RS232 to USB adapter and get the same results.
Any ides what I can do next ?
Thanks, Dick, W1KSZ
Sent from Outlook<http://aka.ms/weboutlook>
BK
Bob kb8tq
Mon, Sep 11, 2017 2:36 AM
Hi
Standard generic drill:
- Swap RX and TX (or check with a voltmeter)
- Cycle through the normal set of baud rates
- Double check the jumpers on the 422 to 232 conversion
- Check with a terminal program, typing in the *IDN? query by hand
Yes it’s sort of a mix and match of what’s above. The order is purely
a “what I usually mess up” sequence.
Bob
On Sep 9, 2017, at 5:24 PM, Richard Solomon w1ksz@outlook.com wrote:
I have a working Z3801A, it has had an AC Supply added to it and the RS422 port has been reconfigured to RS232.
I downloaded and installed Z38XX, configured both the Port and Device to the recommended parameters:
Data Rate: 19,200
Parity: Odd
Data Bits: 7
All I can get, looking at the Device Window is:
Send *IDN ?
Received
Detected Device Undefined
I am assuming the Z801A and the Computer are not talking to each other, but I am at a loss as to what to do next.
I tried a different (known working) RS232 to USB adapter and get the same results.
Any ides what I can do next ?
Thanks, Dick, W1KSZ
Sent from Outlookhttp://aka.ms/weboutlook
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Hi
Standard generic drill:
1) Swap RX and TX (or check with a voltmeter)
2) Cycle through the normal set of baud rates
3) Double check the jumpers on the 422 to 232 conversion
4) Check with a terminal program, typing in the *IDN? query by hand
Yes it’s sort of a mix and match of what’s above. The order is purely
a “what I usually mess up” sequence.
Bob
> On Sep 9, 2017, at 5:24 PM, Richard Solomon <w1ksz@outlook.com> wrote:
>
> I have a working Z3801A, it has had an AC Supply added to it and the RS422 port has been reconfigured to RS232.
>
>
> I downloaded and installed Z38XX, configured both the Port and Device to the recommended parameters:
>
>
> Data Rate: 19,200
>
> Parity: Odd
>
> Data Bits: 7
>
>
> All I can get, looking at the Device Window is:
>
>
> Send *IDN ?
>
> Received
>
> Detected Device Undefined
>
>
> I am assuming the Z801A and the Computer are not talking to each other, but I am at a loss as to what to do next.
>
>
> I tried a different (known working) RS232 to USB adapter and get the same results.
>
>
> Any ides what I can do next ?
>
>
> Thanks, Dick, W1KSZ
>
>
> Sent from Outlook<http://aka.ms/weboutlook>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
BH
Bill Hawkins
Mon, Sep 11, 2017 3:16 AM
Shouldn't that be 8 data bits?
Bill Hawkins
-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Richard
Solomon
Sent: Saturday, September 09, 2017 4:24 PM
I downloaded and installed Z38XX, configured both the Port and Device to
the recommended parameters:
Data Rate: 19,200
Parity: Odd
Data Bits: 7
Shouldn't that be 8 data bits?
Bill Hawkins
-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Richard
Solomon
Sent: Saturday, September 09, 2017 4:24 PM
I downloaded and installed Z38XX, configured both the Port and Device to
the recommended parameters:
Data Rate: 19,200
Parity: Odd
Data Bits: 7
RS
Richard Solomon
Mon, Sep 11, 2017 5:05 AM
Not according to the manual, but then that could be my problem !!
73, Dick, W1KSZ
Sent from Outlookhttp://aka.ms/weboutlook
From: time-nuts time-nuts-bounces@febo.com on behalf of Bill Hawkins bill.iaxs@pobox.com
Sent: Sunday, September 10, 2017 8:16:17 PM
To: 'Discussion of precise time and frequency measurement'
Subject: Re: [time-nuts] Z3801A and Z38XX Issues
Shouldn't that be 8 data bits?
Bill Hawkins
-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Richard
Solomon
Sent: Saturday, September 09, 2017 4:24 PM
I downloaded and installed Z38XX, configured both the Port and Device to
the recommended parameters:
Data Rate: 19,200
Parity: Odd
Data Bits: 7
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Not according to the manual, but then that could be my problem !!
73, Dick, W1KSZ
Sent from Outlook<http://aka.ms/weboutlook>
________________________________
From: time-nuts <time-nuts-bounces@febo.com> on behalf of Bill Hawkins <bill.iaxs@pobox.com>
Sent: Sunday, September 10, 2017 8:16:17 PM
To: 'Discussion of precise time and frequency measurement'
Subject: Re: [time-nuts] Z3801A and Z38XX Issues
Shouldn't that be 8 data bits?
Bill Hawkins
-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Richard
Solomon
Sent: Saturday, September 09, 2017 4:24 PM
I downloaded and installed Z38XX, configured both the Port and Device to
the recommended parameters:
Data Rate: 19,200
Parity: Odd
Data Bits: 7
_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
BS
Brad Stockdale
Sat, Sep 16, 2017 1:27 PM
From what I recall, the manual was accurate on the baud, stop bits,
parity, etc. Just a few thoughts.... Make sure that the serial port is
working as expected... Do a loop back test if possible. Are you using
a USB to Serial adapter? Many of them have some weird quirks. IMHO,
FTDI based adapters work better than others. Forego using the Z38XX.
Try using a plain terminal app and see if you get any weird characters
echo'd back instead of expected responses. Etc... There are a lot of
possible causes when a serial connection isn't working right.
On 9/16/2017 at 12:26 AM, "Richard Solomon" wrote:Not according to
the manual, but then that could be my problem !!
73, Dick, W1KSZ
Sent from Outlook
From: time-nuts on behalf of Bill Hawkins
Sent: Sunday, September 10, 2017 8:16:17 PM
To: 'Discussion of precise time and frequency measurement'
Subject: Re: [time-nuts] Z3801A and Z38XX Issues
Shouldn't that be 8 data bits?
Bill Hawkins
-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of
Richard
Solomon
Sent: Saturday, September 09, 2017 4:24 PM
I downloaded and installed Z38XX, configured both the Port and Device
to
the recommended parameters:
Data Rate: 19,200
Parity: Odd
Data Bits: 7
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
>From what I recall, the manual was accurate on the baud, stop bits,
parity, etc. Just a few thoughts.... Make sure that the serial port is
working as expected... Do a loop back test if possible. Are you using
a USB to Serial adapter? Many of them have some weird quirks. IMHO,
FTDI based adapters work better than others. Forego using the Z38XX.
Try using a plain terminal app and see if you get any weird characters
echo'd back instead of expected responses. Etc... There are a lot of
possible causes when a serial connection isn't working right.
On 9/16/2017 at 12:26 AM, "Richard Solomon" wrote:Not according to
the manual, but then that could be my problem !!
73, Dick, W1KSZ
Sent from Outlook
________________________________
From: time-nuts on behalf of Bill Hawkins
Sent: Sunday, September 10, 2017 8:16:17 PM
To: 'Discussion of precise time and frequency measurement'
Subject: Re: [time-nuts] Z3801A and Z38XX Issues
Shouldn't that be 8 data bits?
Bill Hawkins
-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of
Richard
Solomon
Sent: Saturday, September 09, 2017 4:24 PM
I downloaded and installed Z38XX, configured both the Port and Device
to
the recommended parameters:
Data Rate: 19,200
Parity: Odd
Data Bits: 7
_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
PS
paul swed
Sat, Sep 16, 2017 2:25 PM
Its 7O1 thats how mine has always worked.
Regards
Paul
WB8TSL
On Mon, Sep 11, 2017 at 1:05 AM, Richard Solomon w1ksz@outlook.com wrote:
Not according to the manual, but then that could be my problem !!
73, Dick, W1KSZ
Sent from Outlookhttp://aka.ms/weboutlook
From: time-nuts time-nuts-bounces@febo.com on behalf of Bill Hawkins <
bill.iaxs@pobox.com>
Sent: Sunday, September 10, 2017 8:16:17 PM
To: 'Discussion of precise time and frequency measurement'
Subject: Re: [time-nuts] Z3801A and Z38XX Issues
Shouldn't that be 8 data bits?
Bill Hawkins
-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Richard
Solomon
Sent: Saturday, September 09, 2017 4:24 PM
I downloaded and installed Z38XX, configured both the Port and Device to
the recommended parameters:
Data Rate: 19,200
Parity: Odd
Data Bits: 7
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.
Its 7O1 thats how mine has always worked.
Regards
Paul
WB8TSL
On Mon, Sep 11, 2017 at 1:05 AM, Richard Solomon <w1ksz@outlook.com> wrote:
> Not according to the manual, but then that could be my problem !!
>
>
> 73, Dick, W1KSZ
>
>
> Sent from Outlook<http://aka.ms/weboutlook>
> ________________________________
> From: time-nuts <time-nuts-bounces@febo.com> on behalf of Bill Hawkins <
> bill.iaxs@pobox.com>
> Sent: Sunday, September 10, 2017 8:16:17 PM
> To: 'Discussion of precise time and frequency measurement'
> Subject: Re: [time-nuts] Z3801A and Z38XX Issues
>
> Shouldn't that be 8 data bits?
>
> Bill Hawkins
>
> -----Original Message-----
> From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Richard
> Solomon
> Sent: Saturday, September 09, 2017 4:24 PM
>
> I downloaded and installed Z38XX, configured both the Port and Device to
> the recommended parameters:
>
>
> Data Rate: 19,200
>
> Parity: Odd
>
> Data Bits: 7
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/
> mailman/listinfo/time-nuts
> and follow the instructions there.
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/
> mailman/listinfo/time-nuts
> and follow the instructions there.
>
BK
Bob kb8tq
Sat, Sep 16, 2017 8:22 PM
Hi
The gotcha is that the 3801 is an “ask for it / get it” protocol. If you don’t
get the query formatted correctly, there will be no data coming back. It’s not
like some other GPSDO’s that just yack all the time.
Bob
On Sep 16, 2017, at 9:27 AM, Brad Stockdale taynoo@hush.com wrote:
From what I recall, the manual was accurate on the baud, stop bits,
parity, etc. Just a few thoughts.... Make sure that the serial port is
working as expected... Do a loop back test if possible. Are you using
a USB to Serial adapter? Many of them have some weird quirks. IMHO,
FTDI based adapters work better than others. Forego using the Z38XX.
Try using a plain terminal app and see if you get any weird characters
echo'd back instead of expected responses. Etc... There are a lot of
possible causes when a serial connection isn't working right.
On 9/16/2017 at 12:26 AM, "Richard Solomon" wrote:Not according to
the manual, but then that could be my problem !!
73, Dick, W1KSZ
Sent from Outlook
From: time-nuts on behalf of Bill Hawkins
Sent: Sunday, September 10, 2017 8:16:17 PM
To: 'Discussion of precise time and frequency measurement'
Subject: Re: [time-nuts] Z3801A and Z38XX Issues
Shouldn't that be 8 data bits?
Bill Hawkins
-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of
Richard
Solomon
Sent: Saturday, September 09, 2017 4:24 PM
I downloaded and installed Z38XX, configured both the Port and Device
to
the recommended parameters:
Data Rate: 19,200
Parity: Odd
Data Bits: 7
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Hi
The gotcha is that the 3801 is an “ask for it / get it” protocol. If you don’t
get the query formatted correctly, there will be no data coming back. It’s not
like some other GPSDO’s that just yack all the time.
Bob
> On Sep 16, 2017, at 9:27 AM, Brad Stockdale <taynoo@hush.com> wrote:
>
> From what I recall, the manual was accurate on the baud, stop bits,
> parity, etc. Just a few thoughts.... Make sure that the serial port is
> working as expected... Do a loop back test if possible. Are you using
> a USB to Serial adapter? Many of them have some weird quirks. IMHO,
> FTDI based adapters work better than others. Forego using the Z38XX.
> Try using a plain terminal app and see if you get any weird characters
> echo'd back instead of expected responses. Etc... There are a lot of
> possible causes when a serial connection isn't working right.
>
> On 9/16/2017 at 12:26 AM, "Richard Solomon" wrote:Not according to
> the manual, but then that could be my problem !!
> 73, Dick, W1KSZ
> Sent from Outlook
> ________________________________
> From: time-nuts on behalf of Bill Hawkins
> Sent: Sunday, September 10, 2017 8:16:17 PM
> To: 'Discussion of precise time and frequency measurement'
> Subject: Re: [time-nuts] Z3801A and Z38XX Issues
>
> Shouldn't that be 8 data bits?
>
> Bill Hawkins
>
> -----Original Message-----
> From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of
> Richard
> Solomon
> Sent: Saturday, September 09, 2017 4:24 PM
>
> I downloaded and installed Z38XX, configured both the Port and Device
> to
> the recommended parameters:
> Data Rate: 19,200
>
> Parity: Odd
>
> Data Bits: 7
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
SM
Scott McGrath
Sat, Sep 16, 2017 10:48 PM
Connect to a terminal emulator and IDQ it you will either see garbage in which case serial config is wrong and you need to try a new set of values
Or using a breakout box connect a scope to the txd and rxd lines and compare pulse widths
Content by Scott
Typos by Siri
On Sep 16, 2017, at 4:22 PM, Bob kb8tq kb8tq@n1k.org wrote:
Hi
The gotcha is that the 3801 is an “ask for it / get it” protocol. If you don’t
get the query formatted correctly, there will be no data coming back. It’s not
like some other GPSDO’s that just yack all the time.
Bob
On Sep 16, 2017, at 9:27 AM, Brad Stockdale taynoo@hush.com wrote:
From what I recall, the manual was accurate on the baud, stop bits,
parity, etc. Just a few thoughts.... Make sure that the serial port is
working as expected... Do a loop back test if possible. Are you using
a USB to Serial adapter? Many of them have some weird quirks. IMHO,
FTDI based adapters work better than others. Forego using the Z38XX.
Try using a plain terminal app and see if you get any weird characters
echo'd back instead of expected responses. Etc... There are a lot of
possible causes when a serial connection isn't working right.
On 9/16/2017 at 12:26 AM, "Richard Solomon" wrote:Not according to
the manual, but then that could be my problem !!
73, Dick, W1KSZ
Sent from Outlook
From: time-nuts on behalf of Bill Hawkins
Sent: Sunday, September 10, 2017 8:16:17 PM
To: 'Discussion of precise time and frequency measurement'
Subject: Re: [time-nuts] Z3801A and Z38XX Issues
Shouldn't that be 8 data bits?
Bill Hawkins
-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of
Richard
Solomon
Sent: Saturday, September 09, 2017 4:24 PM
I downloaded and installed Z38XX, configured both the Port and Device
to
the recommended parameters:
Data Rate: 19,200
Parity: Odd
Data Bits: 7
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Connect to a terminal emulator and IDQ it you will either see garbage in which case serial config is wrong and you need to try a new set of values
Or using a breakout box connect a scope to the txd and rxd lines and compare pulse widths
Content by Scott
Typos by Siri
> On Sep 16, 2017, at 4:22 PM, Bob kb8tq <kb8tq@n1k.org> wrote:
>
> Hi
>
> The gotcha is that the 3801 is an “ask for it / get it” protocol. If you don’t
> get the query formatted correctly, there will be no data coming back. It’s not
> like some other GPSDO’s that just yack all the time.
>
> Bob
>
>> On Sep 16, 2017, at 9:27 AM, Brad Stockdale <taynoo@hush.com> wrote:
>>
>> From what I recall, the manual was accurate on the baud, stop bits,
>> parity, etc. Just a few thoughts.... Make sure that the serial port is
>> working as expected... Do a loop back test if possible. Are you using
>> a USB to Serial adapter? Many of them have some weird quirks. IMHO,
>> FTDI based adapters work better than others. Forego using the Z38XX.
>> Try using a plain terminal app and see if you get any weird characters
>> echo'd back instead of expected responses. Etc... There are a lot of
>> possible causes when a serial connection isn't working right.
>>
>> On 9/16/2017 at 12:26 AM, "Richard Solomon" wrote:Not according to
>> the manual, but then that could be my problem !!
>> 73, Dick, W1KSZ
>> Sent from Outlook
>> ________________________________
>> From: time-nuts on behalf of Bill Hawkins
>> Sent: Sunday, September 10, 2017 8:16:17 PM
>> To: 'Discussion of precise time and frequency measurement'
>> Subject: Re: [time-nuts] Z3801A and Z38XX Issues
>>
>> Shouldn't that be 8 data bits?
>>
>> Bill Hawkins
>>
>> -----Original Message-----
>> From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of
>> Richard
>> Solomon
>> Sent: Saturday, September 09, 2017 4:24 PM
>>
>> I downloaded and installed Z38XX, configured both the Port and Device
>> to
>> the recommended parameters:
>> Data Rate: 19,200
>>
>> Parity: Odd
>>
>> Data Bits: 7
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
RS
Richard Solomon
Sun, Sep 17, 2017 2:31 AM
I have a better idea ... anyone want a Z3801A ? I'll swap it for a late model, working
Trimble T-Bolt.
Problem solved.
73, Dick, W1KSZ
Sent from Outlookhttp://aka.ms/weboutlook
From: time-nuts time-nuts-bounces@febo.com on behalf of Scott McGrath scmcgrath@gmail.com
Sent: Saturday, September 16, 2017 3:48:06 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Z3801A and Z38XX Issues
Connect to a terminal emulator and IDQ it you will either see garbage in which case serial config is wrong and you need to try a new set of values
Or using a breakout box connect a scope to the txd and rxd lines and compare pulse widths
Content by Scott
Typos by Siri
On Sep 16, 2017, at 4:22 PM, Bob kb8tq kb8tq@n1k.org wrote:
Hi
The gotcha is that the 3801 is an “ask for it / get it” protocol. If you don’t
get the query formatted correctly, there will be no data coming back. It’s not
like some other GPSDO’s that just yack all the time.
Bob
On Sep 16, 2017, at 9:27 AM, Brad Stockdale taynoo@hush.com wrote:
From what I recall, the manual was accurate on the baud, stop bits,
parity, etc. Just a few thoughts.... Make sure that the serial port is
working as expected... Do a loop back test if possible. Are you using
a USB to Serial adapter? Many of them have some weird quirks. IMHO,
FTDI based adapters work better than others. Forego using the Z38XX.
Try using a plain terminal app and see if you get any weird characters
echo'd back instead of expected responses. Etc... There are a lot of
possible causes when a serial connection isn't working right.
On 9/16/2017 at 12:26 AM, "Richard Solomon" wrote:Not according to
the manual, but then that could be my problem !!
73, Dick, W1KSZ
Sent from Outlook
From: time-nuts on behalf of Bill Hawkins
Sent: Sunday, September 10, 2017 8:16:17 PM
To: 'Discussion of precise time and frequency measurement'
Subject: Re: [time-nuts] Z3801A and Z38XX Issues
Shouldn't that be 8 data bits?
Bill Hawkins
-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of
Richard
Solomon
Sent: Saturday, September 09, 2017 4:24 PM
I downloaded and installed Z38XX, configured both the Port and Device
to
the recommended parameters:
Data Rate: 19,200
Parity: Odd
Data Bits: 7
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
I have a better idea ... anyone want a Z3801A ? I'll swap it for a late model, working
Trimble T-Bolt.
Problem solved.
73, Dick, W1KSZ
Sent from Outlook<http://aka.ms/weboutlook>
________________________________
From: time-nuts <time-nuts-bounces@febo.com> on behalf of Scott McGrath <scmcgrath@gmail.com>
Sent: Saturday, September 16, 2017 3:48:06 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Z3801A and Z38XX Issues
Connect to a terminal emulator and IDQ it you will either see garbage in which case serial config is wrong and you need to try a new set of values
Or using a breakout box connect a scope to the txd and rxd lines and compare pulse widths
Content by Scott
Typos by Siri
> On Sep 16, 2017, at 4:22 PM, Bob kb8tq <kb8tq@n1k.org> wrote:
>
> Hi
>
> The gotcha is that the 3801 is an “ask for it / get it” protocol. If you don’t
> get the query formatted correctly, there will be no data coming back. It’s not
> like some other GPSDO’s that just yack all the time.
>
> Bob
>
>> On Sep 16, 2017, at 9:27 AM, Brad Stockdale <taynoo@hush.com> wrote:
>>
>> From what I recall, the manual was accurate on the baud, stop bits,
>> parity, etc. Just a few thoughts.... Make sure that the serial port is
>> working as expected... Do a loop back test if possible. Are you using
>> a USB to Serial adapter? Many of them have some weird quirks. IMHO,
>> FTDI based adapters work better than others. Forego using the Z38XX.
>> Try using a plain terminal app and see if you get any weird characters
>> echo'd back instead of expected responses. Etc... There are a lot of
>> possible causes when a serial connection isn't working right.
>>
>> On 9/16/2017 at 12:26 AM, "Richard Solomon" wrote:Not according to
>> the manual, but then that could be my problem !!
>> 73, Dick, W1KSZ
>> Sent from Outlook
>> ________________________________
>> From: time-nuts on behalf of Bill Hawkins
>> Sent: Sunday, September 10, 2017 8:16:17 PM
>> To: 'Discussion of precise time and frequency measurement'
>> Subject: Re: [time-nuts] Z3801A and Z38XX Issues
>>
>> Shouldn't that be 8 data bits?
>>
>> Bill Hawkins
>>
>> -----Original Message-----
>> From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of
>> Richard
>> Solomon
>> Sent: Saturday, September 09, 2017 4:24 PM
>>
>> I downloaded and installed Z38XX, configured both the Port and Device
>> to
>> the recommended parameters:
>> Data Rate: 19,200
>>
>> Parity: Odd
>>
>> Data Bits: 7
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
AM
Artek Manuals
Sun, Sep 17, 2017 5:31 AM
You need the null modem adapter which reverses the TX and RX lines on
one end of the cable. Also not all USB to Serial adapters work.. I
bought an old laptop with a real serial port at a junk computer shop
for not much more than USB adapter to solve the RS-232 problem
Dave
On 9/16/2017 6:48 PM, Scott McGrath wrote:
Connect to a terminal emulator and IDQ it you will either see garbage in which case serial config is wrong and you need to try a new set of values
Or using a breakout box connect a scope to the txd and rxd lines and compare pulse widths
Content by Scott
Typos by Siri
On Sep 16, 2017, at 4:22 PM, Bob kb8tq kb8tq@n1k.org wrote:
Hi
The gotcha is that the 3801 is an “ask for it / get it” protocol. If you don’t
get the query formatted correctly, there will be no data coming back. It’s not
like some other GPSDO’s that just yack all the time.
Bob
On Sep 16, 2017, at 9:27 AM, Brad Stockdale taynoo@hush.com wrote:
From what I recall, the manual was accurate on the baud, stop bits,
parity, etc. Just a few thoughts.... Make sure that the serial port is
working as expected... Do a loop back test if possible. Are you using
a USB to Serial adapter? Many of them have some weird quirks. IMHO,
FTDI based adapters work better than others. Forego using the Z38XX.
Try using a plain terminal app and see if you get any weird characters
echo'd back instead of expected responses. Etc... There are a lot of
possible causes when a serial connection isn't working right.
On 9/16/2017 at 12:26 AM, "Richard Solomon" wrote:Not according to
the manual, but then that could be my problem !!
73, Dick, W1KSZ
Sent from Outlook
From: time-nuts on behalf of Bill Hawkins
Sent: Sunday, September 10, 2017 8:16:17 PM
To: 'Discussion of precise time and frequency measurement'
Subject: Re: [time-nuts] Z3801A and Z38XX Issues
Shouldn't that be 8 data bits?
Bill Hawkins
-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of
Richard
Solomon
Sent: Saturday, September 09, 2017 4:24 PM
I downloaded and installed Z38XX, configured both the Port and Device
to
the recommended parameters:
Data Rate: 19,200
Parity: Odd
Data Bits: 7
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
You need the null modem adapter which reverses the TX and RX lines on
one end of the cable. Also not all USB to Serial adapters work.. I
bought an old laptop with a real serial port at a junk computer shop
for not much more than USB adapter to solve the RS-232 problem
Dave
On 9/16/2017 6:48 PM, Scott McGrath wrote:
> Connect to a terminal emulator and IDQ it you will either see garbage in which case serial config is wrong and you need to try a new set of values
>
> Or using a breakout box connect a scope to the txd and rxd lines and compare pulse widths
>
> Content by Scott
> Typos by Siri
>
>> On Sep 16, 2017, at 4:22 PM, Bob kb8tq <kb8tq@n1k.org> wrote:
>>
>> Hi
>>
>> The gotcha is that the 3801 is an “ask for it / get it” protocol. If you don’t
>> get the query formatted correctly, there will be no data coming back. It’s not
>> like some other GPSDO’s that just yack all the time.
>>
>> Bob
>>
>>> On Sep 16, 2017, at 9:27 AM, Brad Stockdale <taynoo@hush.com> wrote:
>>>
>>> From what I recall, the manual was accurate on the baud, stop bits,
>>> parity, etc. Just a few thoughts.... Make sure that the serial port is
>>> working as expected... Do a loop back test if possible. Are you using
>>> a USB to Serial adapter? Many of them have some weird quirks. IMHO,
>>> FTDI based adapters work better than others. Forego using the Z38XX.
>>> Try using a plain terminal app and see if you get any weird characters
>>> echo'd back instead of expected responses. Etc... There are a lot of
>>> possible causes when a serial connection isn't working right.
>>>
>>> On 9/16/2017 at 12:26 AM, "Richard Solomon" wrote:Not according to
>>> the manual, but then that could be my problem !!
>>> 73, Dick, W1KSZ
>>> Sent from Outlook
>>> ________________________________
>>> From: time-nuts on behalf of Bill Hawkins
>>> Sent: Sunday, September 10, 2017 8:16:17 PM
>>> To: 'Discussion of precise time and frequency measurement'
>>> Subject: Re: [time-nuts] Z3801A and Z38XX Issues
>>>
>>> Shouldn't that be 8 data bits?
>>>
>>> Bill Hawkins
>>>
>>> -----Original Message-----
>>> From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of
>>> Richard
>>> Solomon
>>> Sent: Saturday, September 09, 2017 4:24 PM
>>>
>>> I downloaded and installed Z38XX, configured both the Port and Device
>>> to
>>> the recommended parameters:
>>> Data Rate: 19,200
>>>
>>> Parity: Odd
>>>
>>> Data Bits: 7
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts@febo.com
>>> To unsubscribe, go to
>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts@febo.com
>>> To unsubscribe, go to
>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts@febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>> _______________________________________________
>> time-nuts mailing list -- time-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
--
Dave
Manuals@ArtekManuals.com
www.ArtekManuals.com
---
This email has been checked for viruses by Avast antivirus software.
https://www.avast.com/antivirus