BTW, that is the temp sensitivity of the X72 rubidiums that I have tested...


Your OCXO may be happy at 1x10^-11 / C

BTW, that is the temp sensitivity of the X72 rubidiums that I have tested... ---------------------- Your OCXO may be happy at 1x10^-11 / C
BK
Bob kb8tq
Mon, Nov 27, 2017 12:52 AM

Hi

The “gotcha” with compensated devices is that there can be interesting
breakpoints or decoupling to the sensor. Those show up more if the compensation
is working very hard to get the job done. Put another way - if you start at 1x10^-9/C
and compensate to 1x10^-11/C that is very different than starting at 3x10^-11/C.
A pretty good Rb or DOCXO would be a 1x10^-10 -30 to +70 spec. That would get you
into the 1x10^-12/C range. A 1x10^-9 part over the same range would be a good single
oven or a lower spec Rb. One always needs to be careful with those sort of translations.
They may be optimistic by a factor or 3 or 6 or ….

Bob

On Nov 26, 2017, at 7:38 PM, Mark Sims holrum@hotmail.com wrote:

BTW, that is the temp sensitivity of the X72 rubidiums that I have tested...


Your OCXO may be happy at 1x10^-11 / C


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi The “gotcha” with compensated devices is that there can be interesting breakpoints or decoupling to the sensor. Those show up more if the compensation is working very hard to get the job done. Put another way - if you start at 1x10^-9/C and compensate to 1x10^-11/C that is very different than starting at 3x10^-11/C. A pretty good Rb or DOCXO would be a 1x10^-10 -30 to +70 spec. That would get you into the 1x10^-12/C range. A 1x10^-9 part over the same range would be a good single oven or a lower spec Rb. One always needs to be careful with those sort of translations. They may be optimistic by a factor or 3 or 6 or …. Bob > On Nov 26, 2017, at 7:38 PM, Mark Sims <holrum@hotmail.com> wrote: > > BTW, that is the temp sensitivity of the X72 rubidiums that I have tested... > > ---------------------- > > Your OCXO may be happy at 1x10^-11 / C > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.