Obviously its not possible to have large attachments
so resent without the screen photo.
With best regards
Andreas
----- Original Message -----
From: "Andreas Jahn" Andreas_-_Jahn@t-online.de
To: "Randy" randallgrayevans@yahoo.com
Cc: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Saturday, February 04, 2012 5:31 AM
Subject: Re: [volt-nuts] Fluke 731B replacement
I use the attached cirquit:
its an amplifier with 10000 fold amplification and around 0.1 .. 10Hz
bandwidth.
together with a Oscilloscope with 2mV/div you will get 0.2uV/div
capability.
I use 2 standard electrolytic caps (1000uF + 2200 uF)
of good quality which are selected for leakage current
at 10V charge voltage with precharge for a couple of days.
All is powered by 2 * 9V block NiMH batteries.
All within a metal box (the cakes have been taken out).
The noise floor level with 8*Eneloop (around 10V input voltage) is around
0.12uVpp.
(Screen photo is 1 sek/div and 0.2uV/div on my old HAMEG).
So with this cirquit comparative measurements of different references can
be
made.
With best regards
Andreas
----- Original Message -----
From: "Randy" randallgrayevans@yahoo.com
To: "Andreas Jahn" Andreas_-_Jahn@t-online.de
Cc: "Discussion of precise voltage measurement" volt-nuts@febo.com
Sent: Saturday, February 04, 2012 4:17 AM
Subject: Re: [volt-nuts] Fluke 731B replacement
It is the "Thaler" unit. I got it from a cancelled project. Do you have
an
easy test to check the noise level? The Linear Tech app note by Jim
Williams is too expensive for me ($400 input cap).
Randy
Sent from my iPhone
On Feb 3, 2012, at 2:25 PM, "Andreas Jahn" Andreas_-_Jahn@t-online.de
wrote:
Hello Randy,
is your VRE310 a "new" part (branded APEX) or is it a old "Thaler"
reference?
Some months ago I bougth 2 VRE3050A but the measured noise is well above
the typical value in the data sheet.
So if you have a APEX part it would be interesting for me wether the
VRE310 is more quiet in low frequency noise (0.1 .. 10Hz).
For a 10V part it should be about 6uVpp typical.
Do you plan to heat the reference? Or how do you intend to get near to
the
731B specs?
With best regards
Andreas Jahn
----- Original Message ----- From: "Randy Evans"
randallgrayevans@yahoo.com
To: volt-nuts@febo.com
Sent: Friday, February 03, 2012 5:58 PM
Subject: [volt-nuts] Fluke 731B replacement
I am planning on building an equivalent Fluke 731B transfer Standard
using
a VRE310 10V Precision reference and LTC5400 Quad resistor as a precision
10:1 divider, plus ancillary circuits. However, before starting the
detailed design I modeled the Fluke 731B circuit in LTSpice. The 1V,
1.018V and 1.019V outputs are fine but the 0 to 999uV delta E adjustment
circuit for the delta V, 1.018 + delta E and 1.019 + delta E outputs does
not work per the circuit modeled from the schematic. I have double and
triple checked my LTSpice schematic per the Fluke 731B schematic but I
can't see any difference. Maybe I am just too close to the forest to see
the trees.
Note: The Fluke 731B schematic from the manula I have does not show the
value of the delta E adjustment pot but the parts list shows a value of
5K
ohms.
Does anyone have any ideas? I can forward the LTSpice schematic I am
using
to anyone that might be interested, just send me your e-mail address.
Thanks,
Randy Evans
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts