Re: [time-nuts] Anybody want a Thunderbolt power supply?

SS
Scott Stobbe
Fri, Oct 21, 2016 11:44 PM

A little more data on the 7912.

The first plot shows the tempCo of the 7912 measured with ambient
temperature swings "7912_TempCo.png". Which is -150 ppm/degC.

The second plot is off a 7912 logged for an hour or so, "7912_1PLC.png",
nothing too interesting here. However the environmental temperature swing
of about 1 degC/hour is pretty conservative for a DUT sitting in free air.

Finally, an allan devation plot looking at the normalized stability of a
7912 regulator "7912_AllanDeviation.png". Interestingly here, is, how quick
a 15 mK/min temperature swing shoots above the 1/f floor, it's a matter of
seconds.

Now if your PSRR is 1 ppb/V or better, then all of this is comfortably
below the intrinsic noise of a thunderbolt.

On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe scott.j.stobbe@gmail.com
wrote:

Nick had mention that the -12V rail on the thunderbolt has the poorest
PSRR with respect to frequency output, so I first took a look at the
venerable 7912.

The first data-set was taken with a -13.5 VDC input. Attached is the 0.1
Hz to 10 Hz noise of an essentially quiescently loaded 7912, only a 10k
resistor was added as load for preliminary evaluation. With a 60 dB preamp
the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band noise is
15 uVrms, which is about 1.3 ppm rms of the DC mean.

In allan deviation terms, a quiescently loaded 7912 has a spot noise of 7
uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580 ppb/rtHz.
Equivalently speaking, the flicker noise floor of an allan deviation plot
would be sqrt(2*ln(2)) that figure to be 6.8E-7.

Assuming a thunderbolt should be achieving 1/f floor of around 1E-12, it
would need a PSRR of at least 1 ppm/V. I'm sure someone has gone to the
trouble of actually measuring it.

So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't terrible
with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms, only one
order of magnitude off.

The bad side of a 7912 is in long-term stability and tempCo, the sample I
tested had at least a 150 ppm/degC tempCo, which is going to put a serious
lump/bump in the 10s tau to gps crossover point on an allan deviation plot.

On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe scott.j.stobbe@gmail.com
wrote:

I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm
interested to see the 1/f noise of a classic regulator, what load current
do you expect? I can bias a 7805 for the same load and measure the 0.1 to
10 Hz noise.

Also if you have a digital scope without a very good builtin FFT, octave
would be one solution.

On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts <
time-nuts@febo.com> wrote:

Just an update. I’ve built the second prototype board (I skipped over
the first design), and it’s powering my tbolt right now.

The design calls for 15v in (though it would also work with 13.8v). The
+12 output comes from a D2PAK 7812. For +5, there is an AP1509 buck
converter to make around 6.5 volts, then a DPAK 7805. For -12, there is an
MC34063 configured as an inverter to make around -13.75 volts and then a
DPAK 7912.

Steady-state, the system appears to be working just fine. The AP1509’s
inductor and the D2PAK 7812 are just warm to the touch.

I checked for noise and ripple on the outputs and it’s somewhere around
±2 mV or so generally. From what I can see on the scope, there’s no ripple

  • it’s all high frequency noise. I am not absolutely certain that the noise
    measurement represents real noise or the limits of my measuring ability.
    I’m just using the scope probes the scope came with, and 2 mV/div is its
    lowest range.

I haven’t compared the noise with the ex laptop supply that I was using
before, but I’d have to believe it’s cleaner. I don’t really have a way to
check the oscillator’s before and after ADEV. My only other reference is an
FE5680A, and I think the thunderbolt’s going to be far better at lower tau
(where this all matters).

I know also that ±2 mV is still one and perhaps two orders of magnitude
higher than some have called for. But before I attempt to reduce the noise
further, I’d like to know that there are real gains to be had. Would
someone with a Thunderbolt and better output noise measuring wherewithal be
willing to take a prototype and compare it with something that does have µV
levels of noise and ripple so I can get an idea of what there is to gain?
If you like, you can make such comparisons public - no secrets here.

On Aug 30, 2016, at 10:37 PM, Nick Sayer nsayer@kfu.com wrote:

On Aug 30, 2016, at 8:48 PM, Cube Central cubecentral@gmail.com

wrote:

I would be interested, I think.  Planning ahead for if the one I have

for my Thunderbolt fails, I guess.  Are there different models or would a
photo of the input ports on mine be useful?

Actually, what I had in mind is to just put a SIP4 header on the board

for the output and people could wire the “last mile” themselves. The input
is a 2.1mm barrel connector. You use whatever 15W 12VDC wall wart is handy
and plug it right in.

What it really amounts to is that you get +12 volts directly from the

input, then there’s a buck converter to drop the +12 down to +5 and an
inverter to generate -12 from the +12. Those 3 voltages, plus a ground go
to the SIP4.

So it’s just two switching power supplies to turn a +12 volt only

supply into the three-way that the Thunderbolt wants.

It’d be good for around 1500 mA @ 5V and around 50 mA @ -12 (the +12

spec is whatever is left from the source supply’s power spec) - more than
enough for a Thunderbolt. Probably enough for a hard disk or a smallish PC.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

A little more data on the 7912. The first plot shows the tempCo of the 7912 measured with ambient temperature swings "7912_TempCo.png". Which is -150 ppm/degC. The second plot is off a 7912 logged for an hour or so, "7912_1PLC.png", nothing too interesting here. However the environmental temperature swing of about 1 degC/hour is pretty conservative for a DUT sitting in free air. Finally, an allan devation plot looking at the normalized stability of a 7912 regulator "7912_AllanDeviation.png". Interestingly here, is, how quick a 15 mK/min temperature swing shoots above the 1/f floor, it's a matter of seconds. Now if your PSRR is 1 ppb/V or better, then all of this is comfortably below the intrinsic noise of a thunderbolt. On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe <scott.j.stobbe@gmail.com> wrote: > Nick had mention that the -12V rail on the thunderbolt has the poorest > PSRR with respect to frequency output, so I first took a look at the > venerable 7912. > > The first data-set was taken with a -13.5 VDC input. Attached is the 0.1 > Hz to 10 Hz noise of an essentially quiescently loaded 7912, only a 10k > resistor was added as load for preliminary evaluation. With a 60 dB preamp > the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band noise is > 15 uVrms, which is about 1.3 ppm rms of the DC mean. > > In allan deviation terms, a quiescently loaded 7912 has a spot noise of 7 > uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580 ppb/rtHz. > Equivalently speaking, the flicker noise floor of an allan deviation plot > would be sqrt(2*ln(2)) that figure to be 6.8E-7. > > Assuming a thunderbolt should be achieving 1/f floor of around 1E-12, it > would need a PSRR of at least 1 ppm/V. I'm sure someone has gone to the > trouble of actually measuring it. > > So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't terrible > with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms, only one > order of magnitude off. > > The bad side of a 7912 is in long-term stability and tempCo, the sample I > tested had at least a 150 ppm/degC tempCo, which is going to put a serious > lump/bump in the 10s tau to gps crossover point on an allan deviation plot. > > > > On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe <scott.j.stobbe@gmail.com> > wrote: > >> I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm >> interested to see the 1/f noise of a classic regulator, what load current >> do you expect? I can bias a 7805 for the same load and measure the 0.1 to >> 10 Hz noise. >> >> Also if you have a digital scope without a very good builtin FFT, octave >> would be one solution. >> >> On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts < >> time-nuts@febo.com> wrote: >> >>> Just an update. I’ve built the second prototype board (I skipped over >>> the first design), and it’s powering my tbolt right now. >>> >>> The design calls for 15v in (though it would also work with 13.8v). The >>> +12 output comes from a D2PAK 7812. For +5, there is an AP1509 buck >>> converter to make around 6.5 volts, then a DPAK 7805. For -12, there is an >>> MC34063 configured as an inverter to make around -13.75 volts and then a >>> DPAK 7912. >>> >>> Steady-state, the system appears to be working just fine. The AP1509’s >>> inductor and the D2PAK 7812 are just warm to the touch. >>> >>> I checked for noise and ripple on the outputs and it’s somewhere around >>> ±2 mV or so generally. From what I can see on the scope, there’s no ripple >>> - it’s all high frequency noise. I am not absolutely certain that the noise >>> measurement represents real noise or the limits of my measuring ability. >>> I’m just using the scope probes the scope came with, and 2 mV/div is its >>> lowest range. >>> >>> I haven’t compared the noise with the ex laptop supply that I was using >>> before, but I’d have to believe it’s cleaner. I don’t really have a way to >>> check the oscillator’s before and after ADEV. My only other reference is an >>> FE5680A, and I think the thunderbolt’s going to be far better at lower tau >>> (where this all matters). >>> >>> I know also that ±2 mV is still one and perhaps two orders of magnitude >>> higher than some have called for. But before I attempt to reduce the noise >>> further, I’d like to know that there are real gains to be had. Would >>> someone with a Thunderbolt and better output noise measuring wherewithal be >>> willing to take a prototype and compare it with something that does have µV >>> levels of noise and ripple so I can get an idea of what there is to gain? >>> If you like, you can make such comparisons public - no secrets here. >>> >>> > On Aug 30, 2016, at 10:37 PM, Nick Sayer <nsayer@kfu.com> wrote: >>> > >>> > >>> >> On Aug 30, 2016, at 8:48 PM, Cube Central <cubecentral@gmail.com> >>> wrote: >>> >> >>> >> I would be interested, I think. Planning ahead for if the one I have >>> for my Thunderbolt fails, I guess. Are there different models or would a >>> photo of the input ports on mine be useful? >>> > >>> > Actually, what I had in mind is to just put a SIP4 header on the board >>> for the output and people could wire the “last mile” themselves. The input >>> is a 2.1mm barrel connector. You use whatever 15W 12VDC wall wart is handy >>> and plug it right in. >>> > >>> > What it really amounts to is that you get +12 volts directly from the >>> input, then there’s a buck converter to drop the +12 down to +5 and an >>> inverter to generate -12 from the +12. Those 3 voltages, plus a ground go >>> to the SIP4. >>> > >>> > So it’s just two switching power supplies to turn a +12 volt only >>> supply into the three-way that the Thunderbolt wants. >>> > >>> > It’d be good for around 1500 mA @ 5V and around 50 mA @ -12 (the +12 >>> spec is whatever is left from the source supply’s power spec) - more than >>> enough for a Thunderbolt. Probably enough for a hard disk or a smallish PC. >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/m >>> ailman/listinfo/time-nuts >>> and follow the instructions there. >>> >> >> >
BC
Bob Camp
Sat, Oct 22, 2016 12:20 AM

Hi

On Oct 21, 2016, at 7:44 PM, Scott Stobbe scott.j.stobbe@gmail.com wrote:

A little more data on the 7912.

The first plot shows the tempCo of the 7912 measured with ambient
temperature swings "7912_TempCo.png". Which is -150 ppm/degC.

The second plot is off a 7912 logged for an hour or so, "7912_1PLC.png",
nothing too interesting here. However the environmental temperature swing
of about 1 degC/hour is pretty conservative for a DUT sitting in free air.

Finally, an allan devation plot looking at the normalized stability of a
7912 regulator "7912_AllanDeviation.png". Interestingly here, is, how quick
a 15 mK/min temperature swing shoots above the 1/f floor, it's a matter of
seconds.

Now if your PSRR is 1 ppb/V or better, then all of this is comfortably
below the intrinsic noise of a thunderbolt.

The main point is that the internal tempco of the TBolt it’s self is much larger than
the issues surrounding the power supply pins. The +12 is the only one that is
sensitive enough to voltage (change in frequency vs voltage) to contribute to any
significant way to the overall stability.

Bob

On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe scott.j.stobbe@gmail.com
wrote:

Nick had mention that the -12V rail on the thunderbolt has the poorest
PSRR with respect to frequency output, so I first took a look at the
venerable 7912.

The first data-set was taken with a -13.5 VDC input. Attached is the 0.1
Hz to 10 Hz noise of an essentially quiescently loaded 7912, only a 10k
resistor was added as load for preliminary evaluation. With a 60 dB preamp
the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band noise is
15 uVrms, which is about 1.3 ppm rms of the DC mean.

In allan deviation terms, a quiescently loaded 7912 has a spot noise of 7
uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580 ppb/rtHz.
Equivalently speaking, the flicker noise floor of an allan deviation plot
would be sqrt(2*ln(2)) that figure to be 6.8E-7.

Assuming a thunderbolt should be achieving 1/f floor of around 1E-12, it
would need a PSRR of at least 1 ppm/V. I'm sure someone has gone to the
trouble of actually measuring it.

So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't terrible
with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms, only one
order of magnitude off.

The bad side of a 7912 is in long-term stability and tempCo, the sample I
tested had at least a 150 ppm/degC tempCo, which is going to put a serious
lump/bump in the 10s tau to gps crossover point on an allan deviation plot.

On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe scott.j.stobbe@gmail.com
wrote:

I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm
interested to see the 1/f noise of a classic regulator, what load current
do you expect? I can bias a 7805 for the same load and measure the 0.1 to
10 Hz noise.

Also if you have a digital scope without a very good builtin FFT, octave
would be one solution.

On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts <
time-nuts@febo.com> wrote:

Just an update. I’ve built the second prototype board (I skipped over
the first design), and it’s powering my tbolt right now.

The design calls for 15v in (though it would also work with 13.8v). The
+12 output comes from a D2PAK 7812. For +5, there is an AP1509 buck
converter to make around 6.5 volts, then a DPAK 7805. For -12, there is an
MC34063 configured as an inverter to make around -13.75 volts and then a
DPAK 7912.

Steady-state, the system appears to be working just fine. The AP1509’s
inductor and the D2PAK 7812 are just warm to the touch.

I checked for noise and ripple on the outputs and it’s somewhere around
±2 mV or so generally. From what I can see on the scope, there’s no ripple

  • it’s all high frequency noise. I am not absolutely certain that the noise
    measurement represents real noise or the limits of my measuring ability.
    I’m just using the scope probes the scope came with, and 2 mV/div is its
    lowest range.

I haven’t compared the noise with the ex laptop supply that I was using
before, but I’d have to believe it’s cleaner. I don’t really have a way to
check the oscillator’s before and after ADEV. My only other reference is an
FE5680A, and I think the thunderbolt’s going to be far better at lower tau
(where this all matters).

I know also that ±2 mV is still one and perhaps two orders of magnitude
higher than some have called for. But before I attempt to reduce the noise
further, I’d like to know that there are real gains to be had. Would
someone with a Thunderbolt and better output noise measuring wherewithal be
willing to take a prototype and compare it with something that does have µV
levels of noise and ripple so I can get an idea of what there is to gain?
If you like, you can make such comparisons public - no secrets here.

On Aug 30, 2016, at 10:37 PM, Nick Sayer nsayer@kfu.com wrote:

On Aug 30, 2016, at 8:48 PM, Cube Central cubecentral@gmail.com

wrote:

I would be interested, I think.  Planning ahead for if the one I have

for my Thunderbolt fails, I guess.  Are there different models or would a
photo of the input ports on mine be useful?

Actually, what I had in mind is to just put a SIP4 header on the board

for the output and people could wire the “last mile” themselves. The input
is a 2.1mm barrel connector. You use whatever 15W 12VDC wall wart is handy
and plug it right in.

What it really amounts to is that you get +12 volts directly from the

input, then there’s a buck converter to drop the +12 down to +5 and an
inverter to generate -12 from the +12. Those 3 voltages, plus a ground go
to the SIP4.

So it’s just two switching power supplies to turn a +12 volt only

supply into the three-way that the Thunderbolt wants.

It’d be good for around 1500 mA @ 5V and around 50 mA @ -12 (the +12

spec is whatever is left from the source supply’s power spec) - more than
enough for a Thunderbolt. Probably enough for a hard disk or a smallish PC.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

<7912_1PLC.png><7912_AllanDeviation.png><7912_TempCo.png>_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi > On Oct 21, 2016, at 7:44 PM, Scott Stobbe <scott.j.stobbe@gmail.com> wrote: > > A little more data on the 7912. > > The first plot shows the tempCo of the 7912 measured with ambient > temperature swings "7912_TempCo.png". Which is -150 ppm/degC. > > The second plot is off a 7912 logged for an hour or so, "7912_1PLC.png", > nothing too interesting here. However the environmental temperature swing > of about 1 degC/hour is pretty conservative for a DUT sitting in free air. > > Finally, an allan devation plot looking at the normalized stability of a > 7912 regulator "7912_AllanDeviation.png". Interestingly here, is, how quick > a 15 mK/min temperature swing shoots above the 1/f floor, it's a matter of > seconds. > > Now if your PSRR is 1 ppb/V or better, then all of this is comfortably > below the intrinsic noise of a thunderbolt. The main point is that the internal tempco of the TBolt it’s self is much larger than the issues surrounding the power supply pins. The +12 is the only one that is sensitive enough to voltage (change in frequency vs voltage) to contribute to any significant way to the overall stability. Bob > > On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe <scott.j.stobbe@gmail.com> > wrote: > >> Nick had mention that the -12V rail on the thunderbolt has the poorest >> PSRR with respect to frequency output, so I first took a look at the >> venerable 7912. >> >> The first data-set was taken with a -13.5 VDC input. Attached is the 0.1 >> Hz to 10 Hz noise of an essentially quiescently loaded 7912, only a 10k >> resistor was added as load for preliminary evaluation. With a 60 dB preamp >> the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band noise is >> 15 uVrms, which is about 1.3 ppm rms of the DC mean. >> >> In allan deviation terms, a quiescently loaded 7912 has a spot noise of 7 >> uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580 ppb/rtHz. >> Equivalently speaking, the flicker noise floor of an allan deviation plot >> would be sqrt(2*ln(2)) that figure to be 6.8E-7. >> >> Assuming a thunderbolt should be achieving 1/f floor of around 1E-12, it >> would need a PSRR of at least 1 ppm/V. I'm sure someone has gone to the >> trouble of actually measuring it. >> >> So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't terrible >> with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms, only one >> order of magnitude off. >> >> The bad side of a 7912 is in long-term stability and tempCo, the sample I >> tested had at least a 150 ppm/degC tempCo, which is going to put a serious >> lump/bump in the 10s tau to gps crossover point on an allan deviation plot. >> >> >> >> On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe <scott.j.stobbe@gmail.com> >> wrote: >> >>> I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm >>> interested to see the 1/f noise of a classic regulator, what load current >>> do you expect? I can bias a 7805 for the same load and measure the 0.1 to >>> 10 Hz noise. >>> >>> Also if you have a digital scope without a very good builtin FFT, octave >>> would be one solution. >>> >>> On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts < >>> time-nuts@febo.com> wrote: >>> >>>> Just an update. I’ve built the second prototype board (I skipped over >>>> the first design), and it’s powering my tbolt right now. >>>> >>>> The design calls for 15v in (though it would also work with 13.8v). The >>>> +12 output comes from a D2PAK 7812. For +5, there is an AP1509 buck >>>> converter to make around 6.5 volts, then a DPAK 7805. For -12, there is an >>>> MC34063 configured as an inverter to make around -13.75 volts and then a >>>> DPAK 7912. >>>> >>>> Steady-state, the system appears to be working just fine. The AP1509’s >>>> inductor and the D2PAK 7812 are just warm to the touch. >>>> >>>> I checked for noise and ripple on the outputs and it’s somewhere around >>>> ±2 mV or so generally. From what I can see on the scope, there’s no ripple >>>> - it’s all high frequency noise. I am not absolutely certain that the noise >>>> measurement represents real noise or the limits of my measuring ability. >>>> I’m just using the scope probes the scope came with, and 2 mV/div is its >>>> lowest range. >>>> >>>> I haven’t compared the noise with the ex laptop supply that I was using >>>> before, but I’d have to believe it’s cleaner. I don’t really have a way to >>>> check the oscillator’s before and after ADEV. My only other reference is an >>>> FE5680A, and I think the thunderbolt’s going to be far better at lower tau >>>> (where this all matters). >>>> >>>> I know also that ±2 mV is still one and perhaps two orders of magnitude >>>> higher than some have called for. But before I attempt to reduce the noise >>>> further, I’d like to know that there are real gains to be had. Would >>>> someone with a Thunderbolt and better output noise measuring wherewithal be >>>> willing to take a prototype and compare it with something that does have µV >>>> levels of noise and ripple so I can get an idea of what there is to gain? >>>> If you like, you can make such comparisons public - no secrets here. >>>> >>>>> On Aug 30, 2016, at 10:37 PM, Nick Sayer <nsayer@kfu.com> wrote: >>>>> >>>>> >>>>>> On Aug 30, 2016, at 8:48 PM, Cube Central <cubecentral@gmail.com> >>>> wrote: >>>>>> >>>>>> I would be interested, I think. Planning ahead for if the one I have >>>> for my Thunderbolt fails, I guess. Are there different models or would a >>>> photo of the input ports on mine be useful? >>>>> >>>>> Actually, what I had in mind is to just put a SIP4 header on the board >>>> for the output and people could wire the “last mile” themselves. The input >>>> is a 2.1mm barrel connector. You use whatever 15W 12VDC wall wart is handy >>>> and plug it right in. >>>>> >>>>> What it really amounts to is that you get +12 volts directly from the >>>> input, then there’s a buck converter to drop the +12 down to +5 and an >>>> inverter to generate -12 from the +12. Those 3 voltages, plus a ground go >>>> to the SIP4. >>>>> >>>>> So it’s just two switching power supplies to turn a +12 volt only >>>> supply into the three-way that the Thunderbolt wants. >>>>> >>>>> It’d be good for around 1500 mA @ 5V and around 50 mA @ -12 (the +12 >>>> spec is whatever is left from the source supply’s power spec) - more than >>>> enough for a Thunderbolt. Probably enough for a hard disk or a smallish PC. >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/m >>>> ailman/listinfo/time-nuts >>>> and follow the instructions there. >>>> >>> >>> >> > <7912_1PLC.png><7912_AllanDeviation.png><7912_TempCo.png>_______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
SS
Scott Stobbe
Sat, Oct 22, 2016 3:06 AM

Interesting, I would have thought that the +12V input would be extremely
well regulated since its shared with the oven heater, I*R drops are going
to show up every where, if your looking for uV levels of stability.
Just a connector has milliohms of contact resistance, let alone routing and
cables...

On Friday, 21 October 2016, Bob Camp kb8tq@n1k.org wrote:

Hi

On Oct 21, 2016, at 7:44 PM, Scott Stobbe <scott.j.stobbe@gmail.com

javascript:;> wrote:

A little more data on the 7912.

The first plot shows the tempCo of the 7912 measured with ambient
temperature swings "7912_TempCo.png". Which is -150 ppm/degC.

The second plot is off a 7912 logged for an hour or so, "7912_1PLC.png",
nothing too interesting here. However the environmental temperature swing
of about 1 degC/hour is pretty conservative for a DUT sitting in free

air.

Finally, an allan devation plot looking at the normalized stability of a
7912 regulator "7912_AllanDeviation.png". Interestingly here, is, how

quick

a 15 mK/min temperature swing shoots above the 1/f floor, it's a matter

of

seconds.

Now if your PSRR is 1 ppb/V or better, then all of this is comfortably
below the intrinsic noise of a thunderbolt.

The main point is that the internal tempco of the TBolt it’s self is much
larger than
the issues surrounding the power supply pins. The +12 is the only one that
is
sensitive enough to voltage (change in frequency vs voltage) to contribute
to any
significant way to the overall stability.

Bob

On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe <scott.j.stobbe@gmail.com

wrote:

Nick had mention that the -12V rail on the thunderbolt has the poorest
PSRR with respect to frequency output, so I first took a look at the
venerable 7912.

The first data-set was taken with a -13.5 VDC input. Attached is the 0.1
Hz to 10 Hz noise of an essentially quiescently loaded 7912, only a 10k
resistor was added as load for preliminary evaluation. With a 60 dB

preamp

the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band noise

is

15 uVrms, which is about 1.3 ppm rms of the DC mean.

In allan deviation terms, a quiescently loaded 7912 has a spot noise of

7

uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580 ppb/rtHz.
Equivalently speaking, the flicker noise floor of an allan deviation

plot

would be sqrt(2*ln(2)) that figure to be 6.8E-7.

Assuming a thunderbolt should be achieving 1/f floor of around 1E-12, it
would need a PSRR of at least 1 ppm/V. I'm sure someone has gone to the
trouble of actually measuring it.

So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't terrible
with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms, only

one

order of magnitude off.

The bad side of a 7912 is in long-term stability and tempCo, the sample

I

tested had at least a 150 ppm/degC tempCo, which is going to put a

serious

lump/bump in the 10s tau to gps crossover point on an allan deviation

plot.

On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe <scott.j.stobbe@gmail.com

wrote:

I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm
interested to see the 1/f noise of a classic regulator, what load

current

do you expect? I can bias a 7805 for the same load and measure the 0.1

to

10 Hz noise.

Also if you have a digital scope without a very good builtin FFT,

octave

would be one solution.

On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts <
time-nuts@febo.com javascript:;> wrote:

Just an update. I’ve built the second prototype board (I skipped over
the first design), and it’s powering my tbolt right now.

The design calls for 15v in (though it would also work with 13.8v).

The

+12 output comes from a D2PAK 7812. For +5, there is an AP1509 buck
converter to make around 6.5 volts, then a DPAK 7805. For -12, there

is an

MC34063 configured as an inverter to make around -13.75 volts and

then a

DPAK 7912.

Steady-state, the system appears to be working just fine. The AP1509’s
inductor and the D2PAK 7812 are just warm to the touch.

I checked for noise and ripple on the outputs and it’s somewhere

around

±2 mV or so generally. From what I can see on the scope, there’s no

ripple

  • it’s all high frequency noise. I am not absolutely certain that the

noise

measurement represents real noise or the limits of my measuring

ability.

I’m just using the scope probes the scope came with, and 2 mV/div is

its

lowest range.

I haven’t compared the noise with the ex laptop supply that I was

using

before, but I’d have to believe it’s cleaner. I don’t really have a

way to

check the oscillator’s before and after ADEV. My only other reference

is an

FE5680A, and I think the thunderbolt’s going to be far better at

lower tau

(where this all matters).

I know also that ±2 mV is still one and perhaps two orders of

magnitude

higher than some have called for. But before I attempt to reduce the

noise

further, I’d like to know that there are real gains to be had. Would
someone with a Thunderbolt and better output noise measuring

wherewithal be

willing to take a prototype and compare it with something that does

have µV

levels of noise and ripple so I can get an idea of what there is to

gain?

If you like, you can make such comparisons public - no secrets here.

On Aug 30, 2016, at 10:37 PM, Nick Sayer <nsayer@kfu.com

javascript:;> wrote:

On Aug 30, 2016, at 8:48 PM, Cube Central <cubecentral@gmail.com

wrote:

I would be interested, I think.  Planning ahead for if the one I

have

for my Thunderbolt fails, I guess.  Are there different models or

would a

photo of the input ports on mine be useful?

Actually, what I had in mind is to just put a SIP4 header on the

board

for the output and people could wire the “last mile” themselves. The

input

is a 2.1mm barrel connector. You use whatever 15W 12VDC wall wart is

handy

and plug it right in.

What it really amounts to is that you get +12 volts directly from the

input, then there’s a buck converter to drop the +12 down to +5 and an
inverter to generate -12 from the +12. Those 3 voltages, plus a

ground go

to the SIP4.

So it’s just two switching power supplies to turn a +12 volt only

supply into the three-way that the Thunderbolt wants.

It’d be good for around 1500 mA @ 5V and around 50 mA @ -12 (the +12

spec is whatever is left from the source supply’s power spec) - more

than

enough for a Thunderbolt. Probably enough for a hard disk or a

smallish PC.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

<7912_1PLC.png><7912_AllanDeviation.png><7912_

TempCo.png>_______________________________________________

time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

Interesting, I would have thought that the +12V input would be extremely well regulated since its shared with the oven heater, I*R drops are going to show up every where, if your looking for uV levels of stability. Just a connector has milliohms of contact resistance, let alone routing and cables... On Friday, 21 October 2016, Bob Camp <kb8tq@n1k.org> wrote: > Hi > > > On Oct 21, 2016, at 7:44 PM, Scott Stobbe <scott.j.stobbe@gmail.com > <javascript:;>> wrote: > > > > A little more data on the 7912. > > > > The first plot shows the tempCo of the 7912 measured with ambient > > temperature swings "7912_TempCo.png". Which is -150 ppm/degC. > > > > The second plot is off a 7912 logged for an hour or so, "7912_1PLC.png", > > nothing too interesting here. However the environmental temperature swing > > of about 1 degC/hour is pretty conservative for a DUT sitting in free > air. > > > > Finally, an allan devation plot looking at the normalized stability of a > > 7912 regulator "7912_AllanDeviation.png". Interestingly here, is, how > quick > > a 15 mK/min temperature swing shoots above the 1/f floor, it's a matter > of > > seconds. > > > > Now if your PSRR is 1 ppb/V or better, then all of this is comfortably > > below the intrinsic noise of a thunderbolt. > > The main point is that the internal tempco of the TBolt it’s self is much > larger than > the issues surrounding the power supply pins. The +12 is the only one that > is > sensitive enough to voltage (change in frequency vs voltage) to contribute > to any > significant way to the overall stability. > > Bob > > > > > On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe <scott.j.stobbe@gmail.com > <javascript:;>> > > wrote: > > > >> Nick had mention that the -12V rail on the thunderbolt has the poorest > >> PSRR with respect to frequency output, so I first took a look at the > >> venerable 7912. > >> > >> The first data-set was taken with a -13.5 VDC input. Attached is the 0.1 > >> Hz to 10 Hz noise of an essentially quiescently loaded 7912, only a 10k > >> resistor was added as load for preliminary evaluation. With a 60 dB > preamp > >> the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band noise > is > >> 15 uVrms, which is about 1.3 ppm rms of the DC mean. > >> > >> In allan deviation terms, a quiescently loaded 7912 has a spot noise of > 7 > >> uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580 ppb/rtHz. > >> Equivalently speaking, the flicker noise floor of an allan deviation > plot > >> would be sqrt(2*ln(2)) that figure to be 6.8E-7. > >> > >> Assuming a thunderbolt should be achieving 1/f floor of around 1E-12, it > >> would need a PSRR of at least 1 ppm/V. I'm sure someone has gone to the > >> trouble of actually measuring it. > >> > >> So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't terrible > >> with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms, only > one > >> order of magnitude off. > >> > >> The bad side of a 7912 is in long-term stability and tempCo, the sample > I > >> tested had at least a 150 ppm/degC tempCo, which is going to put a > serious > >> lump/bump in the 10s tau to gps crossover point on an allan deviation > plot. > >> > >> > >> > >> On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe <scott.j.stobbe@gmail.com > <javascript:;>> > >> wrote: > >> > >>> I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm > >>> interested to see the 1/f noise of a classic regulator, what load > current > >>> do you expect? I can bias a 7805 for the same load and measure the 0.1 > to > >>> 10 Hz noise. > >>> > >>> Also if you have a digital scope without a very good builtin FFT, > octave > >>> would be one solution. > >>> > >>> On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts < > >>> time-nuts@febo.com <javascript:;>> wrote: > >>> > >>>> Just an update. I’ve built the second prototype board (I skipped over > >>>> the first design), and it’s powering my tbolt right now. > >>>> > >>>> The design calls for 15v in (though it would also work with 13.8v). > The > >>>> +12 output comes from a D2PAK 7812. For +5, there is an AP1509 buck > >>>> converter to make around 6.5 volts, then a DPAK 7805. For -12, there > is an > >>>> MC34063 configured as an inverter to make around -13.75 volts and > then a > >>>> DPAK 7912. > >>>> > >>>> Steady-state, the system appears to be working just fine. The AP1509’s > >>>> inductor and the D2PAK 7812 are just warm to the touch. > >>>> > >>>> I checked for noise and ripple on the outputs and it’s somewhere > around > >>>> ±2 mV or so generally. From what I can see on the scope, there’s no > ripple > >>>> - it’s all high frequency noise. I am not absolutely certain that the > noise > >>>> measurement represents real noise or the limits of my measuring > ability. > >>>> I’m just using the scope probes the scope came with, and 2 mV/div is > its > >>>> lowest range. > >>>> > >>>> I haven’t compared the noise with the ex laptop supply that I was > using > >>>> before, but I’d have to believe it’s cleaner. I don’t really have a > way to > >>>> check the oscillator’s before and after ADEV. My only other reference > is an > >>>> FE5680A, and I think the thunderbolt’s going to be far better at > lower tau > >>>> (where this all matters). > >>>> > >>>> I know also that ±2 mV is still one and perhaps two orders of > magnitude > >>>> higher than some have called for. But before I attempt to reduce the > noise > >>>> further, I’d like to know that there are real gains to be had. Would > >>>> someone with a Thunderbolt and better output noise measuring > wherewithal be > >>>> willing to take a prototype and compare it with something that does > have µV > >>>> levels of noise and ripple so I can get an idea of what there is to > gain? > >>>> If you like, you can make such comparisons public - no secrets here. > >>>> > >>>>> On Aug 30, 2016, at 10:37 PM, Nick Sayer <nsayer@kfu.com > <javascript:;>> wrote: > >>>>> > >>>>> > >>>>>> On Aug 30, 2016, at 8:48 PM, Cube Central <cubecentral@gmail.com > <javascript:;>> > >>>> wrote: > >>>>>> > >>>>>> I would be interested, I think. Planning ahead for if the one I > have > >>>> for my Thunderbolt fails, I guess. Are there different models or > would a > >>>> photo of the input ports on mine be useful? > >>>>> > >>>>> Actually, what I had in mind is to just put a SIP4 header on the > board > >>>> for the output and people could wire the “last mile” themselves. The > input > >>>> is a 2.1mm barrel connector. You use whatever 15W 12VDC wall wart is > handy > >>>> and plug it right in. > >>>>> > >>>>> What it really amounts to is that you get +12 volts directly from the > >>>> input, then there’s a buck converter to drop the +12 down to +5 and an > >>>> inverter to generate -12 from the +12. Those 3 voltages, plus a > ground go > >>>> to the SIP4. > >>>>> > >>>>> So it’s just two switching power supplies to turn a +12 volt only > >>>> supply into the three-way that the Thunderbolt wants. > >>>>> > >>>>> It’d be good for around 1500 mA @ 5V and around 50 mA @ -12 (the +12 > >>>> spec is whatever is left from the source supply’s power spec) - more > than > >>>> enough for a Thunderbolt. Probably enough for a hard disk or a > smallish PC. > >>>> _______________________________________________ > >>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> > >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/m > >>>> ailman/listinfo/time-nuts > >>>> and follow the instructions there. > >>>> > >>> > >>> > >> > > <7912_1PLC.png><7912_AllanDeviation.png><7912_ > TempCo.png>_______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com <javascript:;> > > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com <javascript:;> > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > and follow the instructions there. >
BC
Bob Camp
Sat, Oct 22, 2016 1:06 PM

Hi

The +12 goes straight to the OCXO. All OCXO’s have a voltage sensitivity.
That sensitivity is much higher than voltage sensitivity is much higher than
what you see on the other two supply pins.

Bob

On Oct 21, 2016, at 11:06 PM, Scott Stobbe scott.j.stobbe@gmail.com wrote:

Interesting, I would have thought that the +12V input would be extremely
well regulated since its shared with the oven heater, I*R drops are going
to show up every where, if your looking for uV levels of stability.
Just a connector has milliohms of contact resistance, let alone routing and
cables...

On Friday, 21 October 2016, Bob Camp kb8tq@n1k.org wrote:

Hi

On Oct 21, 2016, at 7:44 PM, Scott Stobbe <scott.j.stobbe@gmail.com

javascript:;> wrote:

A little more data on the 7912.

The first plot shows the tempCo of the 7912 measured with ambient
temperature swings "7912_TempCo.png". Which is -150 ppm/degC.

The second plot is off a 7912 logged for an hour or so, "7912_1PLC.png",
nothing too interesting here. However the environmental temperature swing
of about 1 degC/hour is pretty conservative for a DUT sitting in free

air.

Finally, an allan devation plot looking at the normalized stability of a
7912 regulator "7912_AllanDeviation.png". Interestingly here, is, how

quick

a 15 mK/min temperature swing shoots above the 1/f floor, it's a matter

of

seconds.

Now if your PSRR is 1 ppb/V or better, then all of this is comfortably
below the intrinsic noise of a thunderbolt.

The main point is that the internal tempco of the TBolt it’s self is much
larger than
the issues surrounding the power supply pins. The +12 is the only one that
is
sensitive enough to voltage (change in frequency vs voltage) to contribute
to any
significant way to the overall stability.

Bob

On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe <scott.j.stobbe@gmail.com

wrote:

Nick had mention that the -12V rail on the thunderbolt has the poorest
PSRR with respect to frequency output, so I first took a look at the
venerable 7912.

The first data-set was taken with a -13.5 VDC input. Attached is the 0.1
Hz to 10 Hz noise of an essentially quiescently loaded 7912, only a 10k
resistor was added as load for preliminary evaluation. With a 60 dB

preamp

the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band noise

is

15 uVrms, which is about 1.3 ppm rms of the DC mean.

In allan deviation terms, a quiescently loaded 7912 has a spot noise of

7

uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580 ppb/rtHz.
Equivalently speaking, the flicker noise floor of an allan deviation

plot

would be sqrt(2*ln(2)) that figure to be 6.8E-7.

Assuming a thunderbolt should be achieving 1/f floor of around 1E-12, it
would need a PSRR of at least 1 ppm/V. I'm sure someone has gone to the
trouble of actually measuring it.

So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't terrible
with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms, only

one

order of magnitude off.

The bad side of a 7912 is in long-term stability and tempCo, the sample

I

tested had at least a 150 ppm/degC tempCo, which is going to put a

serious

lump/bump in the 10s tau to gps crossover point on an allan deviation

plot.

On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe <scott.j.stobbe@gmail.com

wrote:

I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm
interested to see the 1/f noise of a classic regulator, what load

current

do you expect? I can bias a 7805 for the same load and measure the 0.1

to

10 Hz noise.

Also if you have a digital scope without a very good builtin FFT,

octave

would be one solution.

On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts <
time-nuts@febo.com javascript:;> wrote:

Just an update. I’ve built the second prototype board (I skipped over
the first design), and it’s powering my tbolt right now.

The design calls for 15v in (though it would also work with 13.8v).

The

+12 output comes from a D2PAK 7812. For +5, there is an AP1509 buck
converter to make around 6.5 volts, then a DPAK 7805. For -12, there

is an

MC34063 configured as an inverter to make around -13.75 volts and

then a

DPAK 7912.

Steady-state, the system appears to be working just fine. The AP1509’s
inductor and the D2PAK 7812 are just warm to the touch.

I checked for noise and ripple on the outputs and it’s somewhere

around

±2 mV or so generally. From what I can see on the scope, there’s no

ripple

  • it’s all high frequency noise. I am not absolutely certain that the

noise

measurement represents real noise or the limits of my measuring

ability.

I’m just using the scope probes the scope came with, and 2 mV/div is

its

lowest range.

I haven’t compared the noise with the ex laptop supply that I was

using

before, but I’d have to believe it’s cleaner. I don’t really have a

way to

check the oscillator’s before and after ADEV. My only other reference

is an

FE5680A, and I think the thunderbolt’s going to be far better at

lower tau

(where this all matters).

I know also that ±2 mV is still one and perhaps two orders of

magnitude

higher than some have called for. But before I attempt to reduce the

noise

further, I’d like to know that there are real gains to be had. Would
someone with a Thunderbolt and better output noise measuring

wherewithal be

willing to take a prototype and compare it with something that does

have µV

levels of noise and ripple so I can get an idea of what there is to

gain?

If you like, you can make such comparisons public - no secrets here.

On Aug 30, 2016, at 10:37 PM, Nick Sayer <nsayer@kfu.com

javascript:;> wrote:

On Aug 30, 2016, at 8:48 PM, Cube Central <cubecentral@gmail.com

wrote:

I would be interested, I think.  Planning ahead for if the one I

have

for my Thunderbolt fails, I guess.  Are there different models or

would a

photo of the input ports on mine be useful?

Actually, what I had in mind is to just put a SIP4 header on the

board

for the output and people could wire the “last mile” themselves. The

input

is a 2.1mm barrel connector. You use whatever 15W 12VDC wall wart is

handy

and plug it right in.

What it really amounts to is that you get +12 volts directly from the

input, then there’s a buck converter to drop the +12 down to +5 and an
inverter to generate -12 from the +12. Those 3 voltages, plus a

ground go

to the SIP4.

So it’s just two switching power supplies to turn a +12 volt only

supply into the three-way that the Thunderbolt wants.

It’d be good for around 1500 mA @ 5V and around 50 mA @ -12 (the +12

spec is whatever is left from the source supply’s power spec) - more

than

enough for a Thunderbolt. Probably enough for a hard disk or a

smallish PC.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

<7912_1PLC.png><7912_AllanDeviation.png><7912_

TempCo.png>_______________________________________________

time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi The +12 goes straight to the OCXO. All OCXO’s have a voltage sensitivity. That sensitivity is much higher than voltage sensitivity is much higher than what you see on the other two supply pins. Bob > On Oct 21, 2016, at 11:06 PM, Scott Stobbe <scott.j.stobbe@gmail.com> wrote: > > Interesting, I would have thought that the +12V input would be extremely > well regulated since its shared with the oven heater, I*R drops are going > to show up every where, if your looking for uV levels of stability. > Just a connector has milliohms of contact resistance, let alone routing and > cables... > > On Friday, 21 October 2016, Bob Camp <kb8tq@n1k.org> wrote: > >> Hi >> >>> On Oct 21, 2016, at 7:44 PM, Scott Stobbe <scott.j.stobbe@gmail.com >> <javascript:;>> wrote: >>> >>> A little more data on the 7912. >>> >>> The first plot shows the tempCo of the 7912 measured with ambient >>> temperature swings "7912_TempCo.png". Which is -150 ppm/degC. >>> >>> The second plot is off a 7912 logged for an hour or so, "7912_1PLC.png", >>> nothing too interesting here. However the environmental temperature swing >>> of about 1 degC/hour is pretty conservative for a DUT sitting in free >> air. >>> >>> Finally, an allan devation plot looking at the normalized stability of a >>> 7912 regulator "7912_AllanDeviation.png". Interestingly here, is, how >> quick >>> a 15 mK/min temperature swing shoots above the 1/f floor, it's a matter >> of >>> seconds. >>> >>> Now if your PSRR is 1 ppb/V or better, then all of this is comfortably >>> below the intrinsic noise of a thunderbolt. >> >> The main point is that the internal tempco of the TBolt it’s self is much >> larger than >> the issues surrounding the power supply pins. The +12 is the only one that >> is >> sensitive enough to voltage (change in frequency vs voltage) to contribute >> to any >> significant way to the overall stability. >> >> Bob >> >>> >>> On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe <scott.j.stobbe@gmail.com >> <javascript:;>> >>> wrote: >>> >>>> Nick had mention that the -12V rail on the thunderbolt has the poorest >>>> PSRR with respect to frequency output, so I first took a look at the >>>> venerable 7912. >>>> >>>> The first data-set was taken with a -13.5 VDC input. Attached is the 0.1 >>>> Hz to 10 Hz noise of an essentially quiescently loaded 7912, only a 10k >>>> resistor was added as load for preliminary evaluation. With a 60 dB >> preamp >>>> the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band noise >> is >>>> 15 uVrms, which is about 1.3 ppm rms of the DC mean. >>>> >>>> In allan deviation terms, a quiescently loaded 7912 has a spot noise of >> 7 >>>> uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580 ppb/rtHz. >>>> Equivalently speaking, the flicker noise floor of an allan deviation >> plot >>>> would be sqrt(2*ln(2)) that figure to be 6.8E-7. >>>> >>>> Assuming a thunderbolt should be achieving 1/f floor of around 1E-12, it >>>> would need a PSRR of at least 1 ppm/V. I'm sure someone has gone to the >>>> trouble of actually measuring it. >>>> >>>> So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't terrible >>>> with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms, only >> one >>>> order of magnitude off. >>>> >>>> The bad side of a 7912 is in long-term stability and tempCo, the sample >> I >>>> tested had at least a 150 ppm/degC tempCo, which is going to put a >> serious >>>> lump/bump in the 10s tau to gps crossover point on an allan deviation >> plot. >>>> >>>> >>>> >>>> On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe <scott.j.stobbe@gmail.com >> <javascript:;>> >>>> wrote: >>>> >>>>> I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm >>>>> interested to see the 1/f noise of a classic regulator, what load >> current >>>>> do you expect? I can bias a 7805 for the same load and measure the 0.1 >> to >>>>> 10 Hz noise. >>>>> >>>>> Also if you have a digital scope without a very good builtin FFT, >> octave >>>>> would be one solution. >>>>> >>>>> On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts < >>>>> time-nuts@febo.com <javascript:;>> wrote: >>>>> >>>>>> Just an update. I’ve built the second prototype board (I skipped over >>>>>> the first design), and it’s powering my tbolt right now. >>>>>> >>>>>> The design calls for 15v in (though it would also work with 13.8v). >> The >>>>>> +12 output comes from a D2PAK 7812. For +5, there is an AP1509 buck >>>>>> converter to make around 6.5 volts, then a DPAK 7805. For -12, there >> is an >>>>>> MC34063 configured as an inverter to make around -13.75 volts and >> then a >>>>>> DPAK 7912. >>>>>> >>>>>> Steady-state, the system appears to be working just fine. The AP1509’s >>>>>> inductor and the D2PAK 7812 are just warm to the touch. >>>>>> >>>>>> I checked for noise and ripple on the outputs and it’s somewhere >> around >>>>>> ±2 mV or so generally. From what I can see on the scope, there’s no >> ripple >>>>>> - it’s all high frequency noise. I am not absolutely certain that the >> noise >>>>>> measurement represents real noise or the limits of my measuring >> ability. >>>>>> I’m just using the scope probes the scope came with, and 2 mV/div is >> its >>>>>> lowest range. >>>>>> >>>>>> I haven’t compared the noise with the ex laptop supply that I was >> using >>>>>> before, but I’d have to believe it’s cleaner. I don’t really have a >> way to >>>>>> check the oscillator’s before and after ADEV. My only other reference >> is an >>>>>> FE5680A, and I think the thunderbolt’s going to be far better at >> lower tau >>>>>> (where this all matters). >>>>>> >>>>>> I know also that ±2 mV is still one and perhaps two orders of >> magnitude >>>>>> higher than some have called for. But before I attempt to reduce the >> noise >>>>>> further, I’d like to know that there are real gains to be had. Would >>>>>> someone with a Thunderbolt and better output noise measuring >> wherewithal be >>>>>> willing to take a prototype and compare it with something that does >> have µV >>>>>> levels of noise and ripple so I can get an idea of what there is to >> gain? >>>>>> If you like, you can make such comparisons public - no secrets here. >>>>>> >>>>>>> On Aug 30, 2016, at 10:37 PM, Nick Sayer <nsayer@kfu.com >> <javascript:;>> wrote: >>>>>>> >>>>>>> >>>>>>>> On Aug 30, 2016, at 8:48 PM, Cube Central <cubecentral@gmail.com >> <javascript:;>> >>>>>> wrote: >>>>>>>> >>>>>>>> I would be interested, I think. Planning ahead for if the one I >> have >>>>>> for my Thunderbolt fails, I guess. Are there different models or >> would a >>>>>> photo of the input ports on mine be useful? >>>>>>> >>>>>>> Actually, what I had in mind is to just put a SIP4 header on the >> board >>>>>> for the output and people could wire the “last mile” themselves. The >> input >>>>>> is a 2.1mm barrel connector. You use whatever 15W 12VDC wall wart is >> handy >>>>>> and plug it right in. >>>>>>> >>>>>>> What it really amounts to is that you get +12 volts directly from the >>>>>> input, then there’s a buck converter to drop the +12 down to +5 and an >>>>>> inverter to generate -12 from the +12. Those 3 voltages, plus a >> ground go >>>>>> to the SIP4. >>>>>>> >>>>>>> So it’s just two switching power supplies to turn a +12 volt only >>>>>> supply into the three-way that the Thunderbolt wants. >>>>>>> >>>>>>> It’d be good for around 1500 mA @ 5V and around 50 mA @ -12 (the +12 >>>>>> spec is whatever is left from the source supply’s power spec) - more >> than >>>>>> enough for a Thunderbolt. Probably enough for a hard disk or a >> smallish PC. >>>>>> _______________________________________________ >>>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/m >>>>>> ailman/listinfo/time-nuts >>>>>> and follow the instructions there. >>>>>> >>>>> >>>>> >>>> >>> <7912_1PLC.png><7912_AllanDeviation.png><7912_ >> TempCo.png>_______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >> mailman/listinfo/time-nuts >>> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >> To unsubscribe, go to https://www.febo.com/cgi-bin/ >> mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
SS
Scott Stobbe
Sat, Oct 22, 2016 2:24 PM

It all depends what the -12V rail is for, some have said it directly
references the EFC dac. I would hope an ocxo would have a better tuning
gain on its efc pin than supply pin but maybe that's not always true.

On Saturday, 22 October 2016, Bob Camp kb8tq@n1k.org wrote:

Hi

The +12 goes straight to the OCXO. All OCXO’s have a voltage sensitivity.
That sensitivity is much higher than voltage sensitivity is much higher
than
what you see on the other two supply pins.

Bob

On Oct 21, 2016, at 11:06 PM, Scott Stobbe <scott.j.stobbe@gmail.com

javascript:;> wrote:

Interesting, I would have thought that the +12V input would be extremely
well regulated since its shared with the oven heater, I*R drops are going
to show up every where, if your looking for uV levels of stability.
Just a connector has milliohms of contact resistance, let alone routing

and

cables...

On Friday, 21 October 2016, Bob Camp <kb8tq@n1k.org javascript:;>

wrote:

Hi

On Oct 21, 2016, at 7:44 PM, Scott Stobbe <scott.j.stobbe@gmail.com

javascript:;> wrote:

A little more data on the 7912.

The first plot shows the tempCo of the 7912 measured with ambient
temperature swings "7912_TempCo.png". Which is -150 ppm/degC.

The second plot is off a 7912 logged for an hour or so,

"7912_1PLC.png",

nothing too interesting here. However the environmental temperature

swing

of about 1 degC/hour is pretty conservative for a DUT sitting in free

air.

Finally, an allan devation plot looking at the normalized stability of

a

7912 regulator "7912_AllanDeviation.png". Interestingly here, is, how

quick

a 15 mK/min temperature swing shoots above the 1/f floor, it's a matter

of

seconds.

Now if your PSRR is 1 ppb/V or better, then all of this is comfortably
below the intrinsic noise of a thunderbolt.

The main point is that the internal tempco of the TBolt it’s self is

much

larger than
the issues surrounding the power supply pins. The +12 is the only one

that

is
sensitive enough to voltage (change in frequency vs voltage) to

contribute

to any
significant way to the overall stability.

Bob

On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe <

wrote:

Nick had mention that the -12V rail on the thunderbolt has the poorest
PSRR with respect to frequency output, so I first took a look at the
venerable 7912.

The first data-set was taken with a -13.5 VDC input. Attached is the

0.1

Hz to 10 Hz noise of an essentially quiescently loaded 7912, only a

10k

resistor was added as load for preliminary evaluation. With a 60 dB

preamp

the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band noise

is

15 uVrms, which is about 1.3 ppm rms of the DC mean.

In allan deviation terms, a quiescently loaded 7912 has a spot noise

of

7

uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580 ppb/rtHz.
Equivalently speaking, the flicker noise floor of an allan deviation

plot

would be sqrt(2*ln(2)) that figure to be 6.8E-7.

Assuming a thunderbolt should be achieving 1/f floor of around 1E-12,

it

would need a PSRR of at least 1 ppm/V. I'm sure someone has gone to

the

trouble of actually measuring it.

So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't terrible
with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms,

only

one

order of magnitude off.

The bad side of a 7912 is in long-term stability and tempCo, the

sample

I

tested had at least a 150 ppm/degC tempCo, which is going to put a

serious

lump/bump in the 10s tau to gps crossover point on an allan deviation

plot.

On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe <

wrote:

I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm
interested to see the 1/f noise of a classic regulator, what load

current

do you expect? I can bias a 7805 for the same load and measure the

0.1

to

10 Hz noise.

Also if you have a digital scope without a very good builtin FFT,

octave

would be one solution.

On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts <
time-nuts@febo.com javascript:; javascript:;> wrote:

Just an update. I’ve built the second prototype board (I skipped

over

the first design), and it’s powering my tbolt right now.

The design calls for 15v in (though it would also work with 13.8v).

The

+12 output comes from a D2PAK 7812. For +5, there is an AP1509 buck
converter to make around 6.5 volts, then a DPAK 7805. For -12, there

is an

MC34063 configured as an inverter to make around -13.75 volts and

then a

DPAK 7912.

Steady-state, the system appears to be working just fine. The

AP1509’s

inductor and the D2PAK 7812 are just warm to the touch.

I checked for noise and ripple on the outputs and it’s somewhere

around

±2 mV or so generally. From what I can see on the scope, there’s no

ripple

  • it’s all high frequency noise. I am not absolutely certain that

the

noise

measurement represents real noise or the limits of my measuring

ability.

I’m just using the scope probes the scope came with, and 2 mV/div is

its

lowest range.

I haven’t compared the noise with the ex laptop supply that I was

using

before, but I’d have to believe it’s cleaner. I don’t really have a

way to

check the oscillator’s before and after ADEV. My only other

reference

is an

FE5680A, and I think the thunderbolt’s going to be far better at

lower tau

(where this all matters).

I know also that ±2 mV is still one and perhaps two orders of

magnitude

higher than some have called for. But before I attempt to reduce the

noise

further, I’d like to know that there are real gains to be had. Would
someone with a Thunderbolt and better output noise measuring

wherewithal be

willing to take a prototype and compare it with something that does

have µV

levels of noise and ripple so I can get an idea of what there is to

gain?

If you like, you can make such comparisons public - no secrets here.

On Aug 30, 2016, at 10:37 PM, Nick Sayer <nsayer@kfu.com

javascript:;> wrote:

On Aug 30, 2016, at 8:48 PM, Cube Central <cubecentral@gmail.com

wrote:

I would be interested, I think.  Planning ahead for if the one I

have

for my Thunderbolt fails, I guess.  Are there different models or

would a

photo of the input ports on mine be useful?

Actually, what I had in mind is to just put a SIP4 header on the

board

for the output and people could wire the “last mile” themselves. The

input

is a 2.1mm barrel connector. You use whatever 15W 12VDC wall wart is

handy

and plug it right in.

What it really amounts to is that you get +12 volts directly from

the

input, then there’s a buck converter to drop the +12 down to +5 and

an

inverter to generate -12 from the +12. Those 3 voltages, plus a

ground go

to the SIP4.

So it’s just two switching power supplies to turn a +12 volt only

supply into the three-way that the Thunderbolt wants.

It’d be good for around 1500 mA @ 5V and around 50 mA @ -12 (the

+12

spec is whatever is left from the source supply’s power spec) - more

than

enough for a Thunderbolt. Probably enough for a hard disk or a

smallish PC.


time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

<7912_1PLC.png><7912_AllanDeviation.png><7912_

TempCo.png>_______________________________________________

time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

It all depends what the -12V rail is for, some have said it directly references the EFC dac. I would hope an ocxo would have a better tuning gain on its efc pin than supply pin but maybe that's not always true. On Saturday, 22 October 2016, Bob Camp <kb8tq@n1k.org> wrote: > Hi > > The +12 goes straight to the OCXO. All OCXO’s have a voltage sensitivity. > That sensitivity is much higher than voltage sensitivity is much higher > than > what you see on the other two supply pins. > > Bob > > > On Oct 21, 2016, at 11:06 PM, Scott Stobbe <scott.j.stobbe@gmail.com > <javascript:;>> wrote: > > > > Interesting, I would have thought that the +12V input would be extremely > > well regulated since its shared with the oven heater, I*R drops are going > > to show up every where, if your looking for uV levels of stability. > > Just a connector has milliohms of contact resistance, let alone routing > and > > cables... > > > > On Friday, 21 October 2016, Bob Camp <kb8tq@n1k.org <javascript:;>> > wrote: > > > >> Hi > >> > >>> On Oct 21, 2016, at 7:44 PM, Scott Stobbe <scott.j.stobbe@gmail.com > <javascript:;> > >> <javascript:;>> wrote: > >>> > >>> A little more data on the 7912. > >>> > >>> The first plot shows the tempCo of the 7912 measured with ambient > >>> temperature swings "7912_TempCo.png". Which is -150 ppm/degC. > >>> > >>> The second plot is off a 7912 logged for an hour or so, > "7912_1PLC.png", > >>> nothing too interesting here. However the environmental temperature > swing > >>> of about 1 degC/hour is pretty conservative for a DUT sitting in free > >> air. > >>> > >>> Finally, an allan devation plot looking at the normalized stability of > a > >>> 7912 regulator "7912_AllanDeviation.png". Interestingly here, is, how > >> quick > >>> a 15 mK/min temperature swing shoots above the 1/f floor, it's a matter > >> of > >>> seconds. > >>> > >>> Now if your PSRR is 1 ppb/V or better, then all of this is comfortably > >>> below the intrinsic noise of a thunderbolt. > >> > >> The main point is that the internal tempco of the TBolt it’s self is > much > >> larger than > >> the issues surrounding the power supply pins. The +12 is the only one > that > >> is > >> sensitive enough to voltage (change in frequency vs voltage) to > contribute > >> to any > >> significant way to the overall stability. > >> > >> Bob > >> > >>> > >>> On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe < > scott.j.stobbe@gmail.com <javascript:;> > >> <javascript:;>> > >>> wrote: > >>> > >>>> Nick had mention that the -12V rail on the thunderbolt has the poorest > >>>> PSRR with respect to frequency output, so I first took a look at the > >>>> venerable 7912. > >>>> > >>>> The first data-set was taken with a -13.5 VDC input. Attached is the > 0.1 > >>>> Hz to 10 Hz noise of an essentially quiescently loaded 7912, only a > 10k > >>>> resistor was added as load for preliminary evaluation. With a 60 dB > >> preamp > >>>> the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band noise > >> is > >>>> 15 uVrms, which is about 1.3 ppm rms of the DC mean. > >>>> > >>>> In allan deviation terms, a quiescently loaded 7912 has a spot noise > of > >> 7 > >>>> uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580 ppb/rtHz. > >>>> Equivalently speaking, the flicker noise floor of an allan deviation > >> plot > >>>> would be sqrt(2*ln(2)) that figure to be 6.8E-7. > >>>> > >>>> Assuming a thunderbolt should be achieving 1/f floor of around 1E-12, > it > >>>> would need a PSRR of at least 1 ppm/V. I'm sure someone has gone to > the > >>>> trouble of actually measuring it. > >>>> > >>>> So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't terrible > >>>> with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms, > only > >> one > >>>> order of magnitude off. > >>>> > >>>> The bad side of a 7912 is in long-term stability and tempCo, the > sample > >> I > >>>> tested had at least a 150 ppm/degC tempCo, which is going to put a > >> serious > >>>> lump/bump in the 10s tau to gps crossover point on an allan deviation > >> plot. > >>>> > >>>> > >>>> > >>>> On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe < > scott.j.stobbe@gmail.com <javascript:;> > >> <javascript:;>> > >>>> wrote: > >>>> > >>>>> I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm > >>>>> interested to see the 1/f noise of a classic regulator, what load > >> current > >>>>> do you expect? I can bias a 7805 for the same load and measure the > 0.1 > >> to > >>>>> 10 Hz noise. > >>>>> > >>>>> Also if you have a digital scope without a very good builtin FFT, > >> octave > >>>>> would be one solution. > >>>>> > >>>>> On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts < > >>>>> time-nuts@febo.com <javascript:;> <javascript:;>> wrote: > >>>>> > >>>>>> Just an update. I’ve built the second prototype board (I skipped > over > >>>>>> the first design), and it’s powering my tbolt right now. > >>>>>> > >>>>>> The design calls for 15v in (though it would also work with 13.8v). > >> The > >>>>>> +12 output comes from a D2PAK 7812. For +5, there is an AP1509 buck > >>>>>> converter to make around 6.5 volts, then a DPAK 7805. For -12, there > >> is an > >>>>>> MC34063 configured as an inverter to make around -13.75 volts and > >> then a > >>>>>> DPAK 7912. > >>>>>> > >>>>>> Steady-state, the system appears to be working just fine. The > AP1509’s > >>>>>> inductor and the D2PAK 7812 are just warm to the touch. > >>>>>> > >>>>>> I checked for noise and ripple on the outputs and it’s somewhere > >> around > >>>>>> ±2 mV or so generally. From what I can see on the scope, there’s no > >> ripple > >>>>>> - it’s all high frequency noise. I am not absolutely certain that > the > >> noise > >>>>>> measurement represents real noise or the limits of my measuring > >> ability. > >>>>>> I’m just using the scope probes the scope came with, and 2 mV/div is > >> its > >>>>>> lowest range. > >>>>>> > >>>>>> I haven’t compared the noise with the ex laptop supply that I was > >> using > >>>>>> before, but I’d have to believe it’s cleaner. I don’t really have a > >> way to > >>>>>> check the oscillator’s before and after ADEV. My only other > reference > >> is an > >>>>>> FE5680A, and I think the thunderbolt’s going to be far better at > >> lower tau > >>>>>> (where this all matters). > >>>>>> > >>>>>> I know also that ±2 mV is still one and perhaps two orders of > >> magnitude > >>>>>> higher than some have called for. But before I attempt to reduce the > >> noise > >>>>>> further, I’d like to know that there are real gains to be had. Would > >>>>>> someone with a Thunderbolt and better output noise measuring > >> wherewithal be > >>>>>> willing to take a prototype and compare it with something that does > >> have µV > >>>>>> levels of noise and ripple so I can get an idea of what there is to > >> gain? > >>>>>> If you like, you can make such comparisons public - no secrets here. > >>>>>> > >>>>>>> On Aug 30, 2016, at 10:37 PM, Nick Sayer <nsayer@kfu.com > <javascript:;> > >> <javascript:;>> wrote: > >>>>>>> > >>>>>>> > >>>>>>>> On Aug 30, 2016, at 8:48 PM, Cube Central <cubecentral@gmail.com > <javascript:;> > >> <javascript:;>> > >>>>>> wrote: > >>>>>>>> > >>>>>>>> I would be interested, I think. Planning ahead for if the one I > >> have > >>>>>> for my Thunderbolt fails, I guess. Are there different models or > >> would a > >>>>>> photo of the input ports on mine be useful? > >>>>>>> > >>>>>>> Actually, what I had in mind is to just put a SIP4 header on the > >> board > >>>>>> for the output and people could wire the “last mile” themselves. The > >> input > >>>>>> is a 2.1mm barrel connector. You use whatever 15W 12VDC wall wart is > >> handy > >>>>>> and plug it right in. > >>>>>>> > >>>>>>> What it really amounts to is that you get +12 volts directly from > the > >>>>>> input, then there’s a buck converter to drop the +12 down to +5 and > an > >>>>>> inverter to generate -12 from the +12. Those 3 voltages, plus a > >> ground go > >>>>>> to the SIP4. > >>>>>>> > >>>>>>> So it’s just two switching power supplies to turn a +12 volt only > >>>>>> supply into the three-way that the Thunderbolt wants. > >>>>>>> > >>>>>>> It’d be good for around 1500 mA @ 5V and around 50 mA @ -12 (the > +12 > >>>>>> spec is whatever is left from the source supply’s power spec) - more > >> than > >>>>>> enough for a Thunderbolt. Probably enough for a hard disk or a > >> smallish PC. > >>>>>> _______________________________________________ > >>>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> > <javascript:;> > >>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/m > >>>>>> ailman/listinfo/time-nuts > >>>>>> and follow the instructions there. > >>>>>> > >>>>> > >>>>> > >>>> > >>> <7912_1PLC.png><7912_AllanDeviation.png><7912_ > >> TempCo.png>_______________________________________________ > >>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> > <javascript:;> > >>> To unsubscribe, go to https://www.febo.com/cgi-bin/ > >> mailman/listinfo/time-nuts > >>> and follow the instructions there. > >> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@febo.com <javascript:;> > <javascript:;> > >> To unsubscribe, go to https://www.febo.com/cgi-bin/ > >> mailman/listinfo/time-nuts > >> and follow the instructions there. > >> > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com <javascript:;> > > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com <javascript:;> > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > and follow the instructions there. >
BC
Bob Camp
Sat, Oct 22, 2016 3:31 PM

Hi

The -12 V line is not the reference for the EFC. It is one supply into the op amp circuit
that drives the EFC. Since the EFC is +/- 5V, there must be both a positive and
a negative supply into the driver circuit for full output swing. The -12 can be anything between about
-13 and -7 without significantly impacting the function of the device. In fact, people
with EFC voltages that are slightly positive have accidentally proven that the
TBolt will work (lock up and function correctly) with zero volts on the -12V line.

Bob

On Oct 22, 2016, at 10:24 AM, Scott Stobbe scott.j.stobbe@gmail.com wrote:

It all depends what the -12V rail is for, some have said it directly
references the EFC dac. I would hope an ocxo would have a better tuning
gain on its efc pin than supply pin but maybe that's not always true.

On Saturday, 22 October 2016, Bob Camp kb8tq@n1k.org wrote:

Hi

The +12 goes straight to the OCXO. All OCXO’s have a voltage sensitivity.
That sensitivity is much higher than voltage sensitivity is much higher
than
what you see on the other two supply pins.

Bob

On Oct 21, 2016, at 11:06 PM, Scott Stobbe <scott.j.stobbe@gmail.com

javascript:;> wrote:

Interesting, I would have thought that the +12V input would be extremely
well regulated since its shared with the oven heater, I*R drops are going
to show up every where, if your looking for uV levels of stability.
Just a connector has milliohms of contact resistance, let alone routing

and

cables...

On Friday, 21 October 2016, Bob Camp <kb8tq@n1k.org javascript:;>

wrote:

Hi

On Oct 21, 2016, at 7:44 PM, Scott Stobbe <scott.j.stobbe@gmail.com

javascript:;> wrote:

A little more data on the 7912.

The first plot shows the tempCo of the 7912 measured with ambient
temperature swings "7912_TempCo.png". Which is -150 ppm/degC.

The second plot is off a 7912 logged for an hour or so,

"7912_1PLC.png",

nothing too interesting here. However the environmental temperature

swing

of about 1 degC/hour is pretty conservative for a DUT sitting in free

air.

Finally, an allan devation plot looking at the normalized stability of

a

7912 regulator "7912_AllanDeviation.png". Interestingly here, is, how

quick

a 15 mK/min temperature swing shoots above the 1/f floor, it's a matter

of

seconds.

Now if your PSRR is 1 ppb/V or better, then all of this is comfortably
below the intrinsic noise of a thunderbolt.

The main point is that the internal tempco of the TBolt it’s self is

much

larger than
the issues surrounding the power supply pins. The +12 is the only one

that

is
sensitive enough to voltage (change in frequency vs voltage) to

contribute

to any
significant way to the overall stability.

Bob

On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe <

wrote:

Nick had mention that the -12V rail on the thunderbolt has the poorest
PSRR with respect to frequency output, so I first took a look at the
venerable 7912.

The first data-set was taken with a -13.5 VDC input. Attached is the

0.1

Hz to 10 Hz noise of an essentially quiescently loaded 7912, only a

10k

resistor was added as load for preliminary evaluation. With a 60 dB

preamp

the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band noise

is

15 uVrms, which is about 1.3 ppm rms of the DC mean.

In allan deviation terms, a quiescently loaded 7912 has a spot noise

of

7

uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580 ppb/rtHz.
Equivalently speaking, the flicker noise floor of an allan deviation

plot

would be sqrt(2*ln(2)) that figure to be 6.8E-7.

Assuming a thunderbolt should be achieving 1/f floor of around 1E-12,

it

would need a PSRR of at least 1 ppm/V. I'm sure someone has gone to

the

trouble of actually measuring it.

So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't terrible
with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms,

only

one

order of magnitude off.

The bad side of a 7912 is in long-term stability and tempCo, the

sample

I

tested had at least a 150 ppm/degC tempCo, which is going to put a

serious

lump/bump in the 10s tau to gps crossover point on an allan deviation

plot.

On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe <

wrote:

I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm
interested to see the 1/f noise of a classic regulator, what load

current

do you expect? I can bias a 7805 for the same load and measure the

0.1

to

10 Hz noise.

Also if you have a digital scope without a very good builtin FFT,

octave

would be one solution.

On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts <
time-nuts@febo.com javascript:; javascript:;> wrote:

Just an update. I’ve built the second prototype board (I skipped

over

the first design), and it’s powering my tbolt right now.

The design calls for 15v in (though it would also work with 13.8v).

The

+12 output comes from a D2PAK 7812. For +5, there is an AP1509 buck
converter to make around 6.5 volts, then a DPAK 7805. For -12, there

is an

MC34063 configured as an inverter to make around -13.75 volts and

then a

DPAK 7912.

Steady-state, the system appears to be working just fine. The

AP1509’s

inductor and the D2PAK 7812 are just warm to the touch.

I checked for noise and ripple on the outputs and it’s somewhere

around

±2 mV or so generally. From what I can see on the scope, there’s no

ripple

  • it’s all high frequency noise. I am not absolutely certain that

the

noise

measurement represents real noise or the limits of my measuring

ability.

I’m just using the scope probes the scope came with, and 2 mV/div is

its

lowest range.

I haven’t compared the noise with the ex laptop supply that I was

using

before, but I’d have to believe it’s cleaner. I don’t really have a

way to

check the oscillator’s before and after ADEV. My only other

reference

is an

FE5680A, and I think the thunderbolt’s going to be far better at

lower tau

(where this all matters).

I know also that ±2 mV is still one and perhaps two orders of

magnitude

higher than some have called for. But before I attempt to reduce the

noise

further, I’d like to know that there are real gains to be had. Would
someone with a Thunderbolt and better output noise measuring

wherewithal be

willing to take a prototype and compare it with something that does

have µV

levels of noise and ripple so I can get an idea of what there is to

gain?

If you like, you can make such comparisons public - no secrets here.

On Aug 30, 2016, at 10:37 PM, Nick Sayer <nsayer@kfu.com

javascript:;> wrote:

On Aug 30, 2016, at 8:48 PM, Cube Central <cubecentral@gmail.com

wrote:

I would be interested, I think.  Planning ahead for if the one I

have

for my Thunderbolt fails, I guess.  Are there different models or

would a

photo of the input ports on mine be useful?

Actually, what I had in mind is to just put a SIP4 header on the

board

for the output and people could wire the “last mile” themselves. The

input

is a 2.1mm barrel connector. You use whatever 15W 12VDC wall wart is

handy

and plug it right in.

What it really amounts to is that you get +12 volts directly from

the

input, then there’s a buck converter to drop the +12 down to +5 and

an

inverter to generate -12 from the +12. Those 3 voltages, plus a

ground go

to the SIP4.

So it’s just two switching power supplies to turn a +12 volt only

supply into the three-way that the Thunderbolt wants.

It’d be good for around 1500 mA @ 5V and around 50 mA @ -12 (the

+12

spec is whatever is left from the source supply’s power spec) - more

than

enough for a Thunderbolt. Probably enough for a hard disk or a

smallish PC.


time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

<7912_1PLC.png><7912_AllanDeviation.png><7912_

TempCo.png>_______________________________________________

time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi The -12 V line is not the reference for the EFC. It *is* one supply into the op amp circuit that drives the EFC. Since the EFC is +/- 5V, there must be both a positive and a negative supply into the driver circuit for full output swing. The -12 can be anything between about -13 and -7 without significantly impacting the function of the device. In fact, people with EFC voltages that are slightly positive have accidentally proven that the TBolt will work (lock up and function correctly) with zero volts on the -12V line. Bob > On Oct 22, 2016, at 10:24 AM, Scott Stobbe <scott.j.stobbe@gmail.com> wrote: > > It all depends what the -12V rail is for, some have said it directly > references the EFC dac. I would hope an ocxo would have a better tuning > gain on its efc pin than supply pin but maybe that's not always true. > > On Saturday, 22 October 2016, Bob Camp <kb8tq@n1k.org> wrote: > >> Hi >> >> The +12 goes straight to the OCXO. All OCXO’s have a voltage sensitivity. >> That sensitivity is much higher than voltage sensitivity is much higher >> than >> what you see on the other two supply pins. >> >> Bob >> >>> On Oct 21, 2016, at 11:06 PM, Scott Stobbe <scott.j.stobbe@gmail.com >> <javascript:;>> wrote: >>> >>> Interesting, I would have thought that the +12V input would be extremely >>> well regulated since its shared with the oven heater, I*R drops are going >>> to show up every where, if your looking for uV levels of stability. >>> Just a connector has milliohms of contact resistance, let alone routing >> and >>> cables... >>> >>> On Friday, 21 October 2016, Bob Camp <kb8tq@n1k.org <javascript:;>> >> wrote: >>> >>>> Hi >>>> >>>>> On Oct 21, 2016, at 7:44 PM, Scott Stobbe <scott.j.stobbe@gmail.com >> <javascript:;> >>>> <javascript:;>> wrote: >>>>> >>>>> A little more data on the 7912. >>>>> >>>>> The first plot shows the tempCo of the 7912 measured with ambient >>>>> temperature swings "7912_TempCo.png". Which is -150 ppm/degC. >>>>> >>>>> The second plot is off a 7912 logged for an hour or so, >> "7912_1PLC.png", >>>>> nothing too interesting here. However the environmental temperature >> swing >>>>> of about 1 degC/hour is pretty conservative for a DUT sitting in free >>>> air. >>>>> >>>>> Finally, an allan devation plot looking at the normalized stability of >> a >>>>> 7912 regulator "7912_AllanDeviation.png". Interestingly here, is, how >>>> quick >>>>> a 15 mK/min temperature swing shoots above the 1/f floor, it's a matter >>>> of >>>>> seconds. >>>>> >>>>> Now if your PSRR is 1 ppb/V or better, then all of this is comfortably >>>>> below the intrinsic noise of a thunderbolt. >>>> >>>> The main point is that the internal tempco of the TBolt it’s self is >> much >>>> larger than >>>> the issues surrounding the power supply pins. The +12 is the only one >> that >>>> is >>>> sensitive enough to voltage (change in frequency vs voltage) to >> contribute >>>> to any >>>> significant way to the overall stability. >>>> >>>> Bob >>>> >>>>> >>>>> On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe < >> scott.j.stobbe@gmail.com <javascript:;> >>>> <javascript:;>> >>>>> wrote: >>>>> >>>>>> Nick had mention that the -12V rail on the thunderbolt has the poorest >>>>>> PSRR with respect to frequency output, so I first took a look at the >>>>>> venerable 7912. >>>>>> >>>>>> The first data-set was taken with a -13.5 VDC input. Attached is the >> 0.1 >>>>>> Hz to 10 Hz noise of an essentially quiescently loaded 7912, only a >> 10k >>>>>> resistor was added as load for preliminary evaluation. With a 60 dB >>>> preamp >>>>>> the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band noise >>>> is >>>>>> 15 uVrms, which is about 1.3 ppm rms of the DC mean. >>>>>> >>>>>> In allan deviation terms, a quiescently loaded 7912 has a spot noise >> of >>>> 7 >>>>>> uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580 ppb/rtHz. >>>>>> Equivalently speaking, the flicker noise floor of an allan deviation >>>> plot >>>>>> would be sqrt(2*ln(2)) that figure to be 6.8E-7. >>>>>> >>>>>> Assuming a thunderbolt should be achieving 1/f floor of around 1E-12, >> it >>>>>> would need a PSRR of at least 1 ppm/V. I'm sure someone has gone to >> the >>>>>> trouble of actually measuring it. >>>>>> >>>>>> So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't terrible >>>>>> with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms, >> only >>>> one >>>>>> order of magnitude off. >>>>>> >>>>>> The bad side of a 7912 is in long-term stability and tempCo, the >> sample >>>> I >>>>>> tested had at least a 150 ppm/degC tempCo, which is going to put a >>>> serious >>>>>> lump/bump in the 10s tau to gps crossover point on an allan deviation >>>> plot. >>>>>> >>>>>> >>>>>> >>>>>> On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe < >> scott.j.stobbe@gmail.com <javascript:;> >>>> <javascript:;>> >>>>>> wrote: >>>>>> >>>>>>> I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm >>>>>>> interested to see the 1/f noise of a classic regulator, what load >>>> current >>>>>>> do you expect? I can bias a 7805 for the same load and measure the >> 0.1 >>>> to >>>>>>> 10 Hz noise. >>>>>>> >>>>>>> Also if you have a digital scope without a very good builtin FFT, >>>> octave >>>>>>> would be one solution. >>>>>>> >>>>>>> On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts < >>>>>>> time-nuts@febo.com <javascript:;> <javascript:;>> wrote: >>>>>>> >>>>>>>> Just an update. I’ve built the second prototype board (I skipped >> over >>>>>>>> the first design), and it’s powering my tbolt right now. >>>>>>>> >>>>>>>> The design calls for 15v in (though it would also work with 13.8v). >>>> The >>>>>>>> +12 output comes from a D2PAK 7812. For +5, there is an AP1509 buck >>>>>>>> converter to make around 6.5 volts, then a DPAK 7805. For -12, there >>>> is an >>>>>>>> MC34063 configured as an inverter to make around -13.75 volts and >>>> then a >>>>>>>> DPAK 7912. >>>>>>>> >>>>>>>> Steady-state, the system appears to be working just fine. The >> AP1509’s >>>>>>>> inductor and the D2PAK 7812 are just warm to the touch. >>>>>>>> >>>>>>>> I checked for noise and ripple on the outputs and it’s somewhere >>>> around >>>>>>>> ±2 mV or so generally. From what I can see on the scope, there’s no >>>> ripple >>>>>>>> - it’s all high frequency noise. I am not absolutely certain that >> the >>>> noise >>>>>>>> measurement represents real noise or the limits of my measuring >>>> ability. >>>>>>>> I’m just using the scope probes the scope came with, and 2 mV/div is >>>> its >>>>>>>> lowest range. >>>>>>>> >>>>>>>> I haven’t compared the noise with the ex laptop supply that I was >>>> using >>>>>>>> before, but I’d have to believe it’s cleaner. I don’t really have a >>>> way to >>>>>>>> check the oscillator’s before and after ADEV. My only other >> reference >>>> is an >>>>>>>> FE5680A, and I think the thunderbolt’s going to be far better at >>>> lower tau >>>>>>>> (where this all matters). >>>>>>>> >>>>>>>> I know also that ±2 mV is still one and perhaps two orders of >>>> magnitude >>>>>>>> higher than some have called for. But before I attempt to reduce the >>>> noise >>>>>>>> further, I’d like to know that there are real gains to be had. Would >>>>>>>> someone with a Thunderbolt and better output noise measuring >>>> wherewithal be >>>>>>>> willing to take a prototype and compare it with something that does >>>> have µV >>>>>>>> levels of noise and ripple so I can get an idea of what there is to >>>> gain? >>>>>>>> If you like, you can make such comparisons public - no secrets here. >>>>>>>> >>>>>>>>> On Aug 30, 2016, at 10:37 PM, Nick Sayer <nsayer@kfu.com >> <javascript:;> >>>> <javascript:;>> wrote: >>>>>>>>> >>>>>>>>> >>>>>>>>>> On Aug 30, 2016, at 8:48 PM, Cube Central <cubecentral@gmail.com >> <javascript:;> >>>> <javascript:;>> >>>>>>>> wrote: >>>>>>>>>> >>>>>>>>>> I would be interested, I think. Planning ahead for if the one I >>>> have >>>>>>>> for my Thunderbolt fails, I guess. Are there different models or >>>> would a >>>>>>>> photo of the input ports on mine be useful? >>>>>>>>> >>>>>>>>> Actually, what I had in mind is to just put a SIP4 header on the >>>> board >>>>>>>> for the output and people could wire the “last mile” themselves. The >>>> input >>>>>>>> is a 2.1mm barrel connector. You use whatever 15W 12VDC wall wart is >>>> handy >>>>>>>> and plug it right in. >>>>>>>>> >>>>>>>>> What it really amounts to is that you get +12 volts directly from >> the >>>>>>>> input, then there’s a buck converter to drop the +12 down to +5 and >> an >>>>>>>> inverter to generate -12 from the +12. Those 3 voltages, plus a >>>> ground go >>>>>>>> to the SIP4. >>>>>>>>> >>>>>>>>> So it’s just two switching power supplies to turn a +12 volt only >>>>>>>> supply into the three-way that the Thunderbolt wants. >>>>>>>>> >>>>>>>>> It’d be good for around 1500 mA @ 5V and around 50 mA @ -12 (the >> +12 >>>>>>>> spec is whatever is left from the source supply’s power spec) - more >>>> than >>>>>>>> enough for a Thunderbolt. Probably enough for a hard disk or a >>>> smallish PC. >>>>>>>> _______________________________________________ >>>>>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >> <javascript:;> >>>>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/m >>>>>>>> ailman/listinfo/time-nuts >>>>>>>> and follow the instructions there. >>>>>>>> >>>>>>> >>>>>>> >>>>>> >>>>> <7912_1PLC.png><7912_AllanDeviation.png><7912_ >>>> TempCo.png>_______________________________________________ >>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >> <javascript:;> >>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>>> mailman/listinfo/time-nuts >>>>> and follow the instructions there. >>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >> <javascript:;> >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>>> mailman/listinfo/time-nuts >>>> and follow the instructions there. >>>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >> mailman/listinfo/time-nuts >>> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >> To unsubscribe, go to https://www.febo.com/cgi-bin/ >> mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
SS
Scott Stobbe
Sat, Oct 22, 2016 4:41 PM

Bob, this is good data and insights thanks for taking the time to share.
Ultimately the question Nick is hoping to answer is what is the point of
diminishing returns for voltage regulation. I think there are plenty of
folks on this list that have shared data suggesting a switching regulator
which meets these specifications:

Table 3-2 Power Specifications (Board-only)
+12v +/-10% max current 750 mA
+5v +/-5% max current 400 mA
-12v +/-10% max current 10 mA
Ripple
+5v 50 mV peak-to-peak, 15 mVrms
+/-12v 75 mV peak-to-peak, 20 mVrms

definitely has a degrading impact on the 10MHz phase noise.

I would happily agree that PSRR of an opamp buffering the EFC line is not
going to have a significant impact on performance with a jellybean LDO.

Do you know if the +- 5V reference is derived on board or is it from an EFC
reference provided by the ocxo?

On Sat, Oct 22, 2016 at 11:31 AM, Bob Camp kb8tq@n1k.org wrote:

Hi

The -12 V line is not the reference for the EFC. It is one supply into
the op amp circuit
that drives the EFC. Since the EFC is +/- 5V, there must be both a
positive and
a negative supply into the driver circuit for full output swing. The -12
can be anything between about
-13 and -7 without significantly impacting the function of the device. In
fact, people
with EFC voltages that are slightly positive have accidentally proven that
the
TBolt will work (lock up and function correctly) with zero volts on the
-12V line.

Bob

On Oct 22, 2016, at 10:24 AM, Scott Stobbe scott.j.stobbe@gmail.com

wrote:

It all depends what the -12V rail is for, some have said it directly
references the EFC dac. I would hope an ocxo would have a better tuning
gain on its efc pin than supply pin but maybe that's not always true.

On Saturday, 22 October 2016, Bob Camp kb8tq@n1k.org wrote:

Hi

The +12 goes straight to the OCXO. All OCXO’s have a voltage

sensitivity.

That sensitivity is much higher than voltage sensitivity is much higher
than
what you see on the other two supply pins.

Bob

On Oct 21, 2016, at 11:06 PM, Scott Stobbe <scott.j.stobbe@gmail.com

javascript:;> wrote:

Interesting, I would have thought that the +12V input would be

extremely

well regulated since its shared with the oven heater, I*R drops are

going

to show up every where, if your looking for uV levels of stability.
Just a connector has milliohms of contact resistance, let alone routing

and

cables...

On Friday, 21 October 2016, Bob Camp <kb8tq@n1k.org javascript:;>

wrote:

Hi

On Oct 21, 2016, at 7:44 PM, Scott Stobbe <scott.j.stobbe@gmail.com

javascript:;> wrote:

A little more data on the 7912.

The first plot shows the tempCo of the 7912 measured with ambient
temperature swings "7912_TempCo.png". Which is -150 ppm/degC.

The second plot is off a 7912 logged for an hour or so,

"7912_1PLC.png",

nothing too interesting here. However the environmental temperature

swing

of about 1 degC/hour is pretty conservative for a DUT sitting in free

air.

Finally, an allan devation plot looking at the normalized stability

of

a

7912 regulator "7912_AllanDeviation.png". Interestingly here, is, how

quick

a 15 mK/min temperature swing shoots above the 1/f floor, it's a

matter

of

seconds.

Now if your PSRR is 1 ppb/V or better, then all of this is

comfortably

below the intrinsic noise of a thunderbolt.

The main point is that the internal tempco of the TBolt it’s self is

much

larger than
the issues surrounding the power supply pins. The +12 is the only one

that

is
sensitive enough to voltage (change in frequency vs voltage) to

contribute

to any
significant way to the overall stability.

Bob

On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe <

wrote:

Nick had mention that the -12V rail on the thunderbolt has the

poorest

PSRR with respect to frequency output, so I first took a look at the
venerable 7912.

The first data-set was taken with a -13.5 VDC input. Attached is the

0.1

Hz to 10 Hz noise of an essentially quiescently loaded 7912, only a

10k

resistor was added as load for preliminary evaluation. With a 60 dB

preamp

the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band

noise

is

15 uVrms, which is about 1.3 ppm rms of the DC mean.

In allan deviation terms, a quiescently loaded 7912 has a spot noise

of

7

uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580 ppb/rtHz.
Equivalently speaking, the flicker noise floor of an allan deviation

plot

would be sqrt(2*ln(2)) that figure to be 6.8E-7.

Assuming a thunderbolt should be achieving 1/f floor of around

1E-12,

it

would need a PSRR of at least 1 ppm/V. I'm sure someone has gone to

the

trouble of actually measuring it.

So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't terrible
with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms,

only

one

order of magnitude off.

The bad side of a 7912 is in long-term stability and tempCo, the

sample

I

tested had at least a 150 ppm/degC tempCo, which is going to put a

serious

lump/bump in the 10s tau to gps crossover point on an allan

deviation

plot.

On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe <

wrote:

I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm
interested to see the 1/f noise of a classic regulator, what load

current

do you expect? I can bias a 7805 for the same load and measure the

0.1

to

10 Hz noise.

Also if you have a digital scope without a very good builtin FFT,

octave

would be one solution.

On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts <
time-nuts@febo.com javascript:; javascript:;> wrote:

Just an update. I’ve built the second prototype board (I skipped

over

the first design), and it’s powering my tbolt right now.

The design calls for 15v in (though it would also work with

13.8v).

The

+12 output comes from a D2PAK 7812. For +5, there is an AP1509

buck

converter to make around 6.5 volts, then a DPAK 7805. For -12,

there

is an

MC34063 configured as an inverter to make around -13.75 volts and

then a

DPAK 7912.

Steady-state, the system appears to be working just fine. The

AP1509’s

inductor and the D2PAK 7812 are just warm to the touch.

I checked for noise and ripple on the outputs and it’s somewhere

around

±2 mV or so generally. From what I can see on the scope, there’s

no

ripple

  • it’s all high frequency noise. I am not absolutely certain that

the

noise

measurement represents real noise or the limits of my measuring

ability.

I’m just using the scope probes the scope came with, and 2 mV/div

is

its

lowest range.

I haven’t compared the noise with the ex laptop supply that I was

using

before, but I’d have to believe it’s cleaner. I don’t really have

a

way to

check the oscillator’s before and after ADEV. My only other

reference

is an

FE5680A, and I think the thunderbolt’s going to be far better at

lower tau

(where this all matters).

I know also that ±2 mV is still one and perhaps two orders of

magnitude

higher than some have called for. But before I attempt to reduce

the

noise

further, I’d like to know that there are real gains to be had.

Would

someone with a Thunderbolt and better output noise measuring

wherewithal be

willing to take a prototype and compare it with something that

does

have µV

levels of noise and ripple so I can get an idea of what there is

to

gain?

If you like, you can make such comparisons public - no secrets

here.

On Aug 30, 2016, at 10:37 PM, Nick Sayer <nsayer@kfu.com

javascript:;> wrote:

On Aug 30, 2016, at 8:48 PM, Cube Central <

wrote:

I would be interested, I think.  Planning ahead for if the one I

have

for my Thunderbolt fails, I guess.  Are there different models or

would a

photo of the input ports on mine be useful?

Actually, what I had in mind is to just put a SIP4 header on the

board

for the output and people could wire the “last mile” themselves.

The

input

is a 2.1mm barrel connector. You use whatever 15W 12VDC wall wart

is

handy

and plug it right in.

What it really amounts to is that you get +12 volts directly from

the

input, then there’s a buck converter to drop the +12 down to +5

and

an

inverter to generate -12 from the +12. Those 3 voltages, plus a

ground go

to the SIP4.

So it’s just two switching power supplies to turn a +12 volt only

supply into the three-way that the Thunderbolt wants.

It’d be good for around 1500 mA @ 5V and around 50 mA @ -12 (the

+12

spec is whatever is left from the source supply’s power spec) -

more

than

enough for a Thunderbolt. Probably enough for a hard disk or a

smallish PC.


time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

<7912_1PLC.png><7912_AllanDeviation.png><7912_

TempCo.png>_______________________________________________

time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

Bob, this is good data and insights thanks for taking the time to share. Ultimately the question Nick is hoping to answer is what is the point of diminishing returns for voltage regulation. I think there are plenty of folks on this list that have shared data suggesting a switching regulator which meets these specifications: Table 3-2 Power Specifications (Board-only) +12v +/-10% max current 750 mA +5v +/-5% max current 400 mA -12v +/-10% max current 10 mA Ripple +5v 50 mV peak-to-peak, 15 mVrms +/-12v 75 mV peak-to-peak, 20 mVrms definitely has a degrading impact on the 10MHz phase noise. I would happily agree that PSRR of an opamp buffering the EFC line is not going to have a significant impact on performance with a jellybean LDO. Do you know if the +- 5V reference is derived on board or is it from an EFC reference provided by the ocxo? On Sat, Oct 22, 2016 at 11:31 AM, Bob Camp <kb8tq@n1k.org> wrote: > Hi > > The -12 V line is not the reference for the EFC. It *is* one supply into > the op amp circuit > that drives the EFC. Since the EFC is +/- 5V, there must be both a > positive and > a negative supply into the driver circuit for full output swing. The -12 > can be anything between about > -13 and -7 without significantly impacting the function of the device. In > fact, people > with EFC voltages that are slightly positive have accidentally proven that > the > TBolt will work (lock up and function correctly) with zero volts on the > -12V line. > > Bob > > > On Oct 22, 2016, at 10:24 AM, Scott Stobbe <scott.j.stobbe@gmail.com> > wrote: > > > > It all depends what the -12V rail is for, some have said it directly > > references the EFC dac. I would hope an ocxo would have a better tuning > > gain on its efc pin than supply pin but maybe that's not always true. > > > > On Saturday, 22 October 2016, Bob Camp <kb8tq@n1k.org> wrote: > > > >> Hi > >> > >> The +12 goes straight to the OCXO. All OCXO’s have a voltage > sensitivity. > >> That sensitivity is much higher than voltage sensitivity is much higher > >> than > >> what you see on the other two supply pins. > >> > >> Bob > >> > >>> On Oct 21, 2016, at 11:06 PM, Scott Stobbe <scott.j.stobbe@gmail.com > >> <javascript:;>> wrote: > >>> > >>> Interesting, I would have thought that the +12V input would be > extremely > >>> well regulated since its shared with the oven heater, I*R drops are > going > >>> to show up every where, if your looking for uV levels of stability. > >>> Just a connector has milliohms of contact resistance, let alone routing > >> and > >>> cables... > >>> > >>> On Friday, 21 October 2016, Bob Camp <kb8tq@n1k.org <javascript:;>> > >> wrote: > >>> > >>>> Hi > >>>> > >>>>> On Oct 21, 2016, at 7:44 PM, Scott Stobbe <scott.j.stobbe@gmail.com > >> <javascript:;> > >>>> <javascript:;>> wrote: > >>>>> > >>>>> A little more data on the 7912. > >>>>> > >>>>> The first plot shows the tempCo of the 7912 measured with ambient > >>>>> temperature swings "7912_TempCo.png". Which is -150 ppm/degC. > >>>>> > >>>>> The second plot is off a 7912 logged for an hour or so, > >> "7912_1PLC.png", > >>>>> nothing too interesting here. However the environmental temperature > >> swing > >>>>> of about 1 degC/hour is pretty conservative for a DUT sitting in free > >>>> air. > >>>>> > >>>>> Finally, an allan devation plot looking at the normalized stability > of > >> a > >>>>> 7912 regulator "7912_AllanDeviation.png". Interestingly here, is, how > >>>> quick > >>>>> a 15 mK/min temperature swing shoots above the 1/f floor, it's a > matter > >>>> of > >>>>> seconds. > >>>>> > >>>>> Now if your PSRR is 1 ppb/V or better, then all of this is > comfortably > >>>>> below the intrinsic noise of a thunderbolt. > >>>> > >>>> The main point is that the internal tempco of the TBolt it’s self is > >> much > >>>> larger than > >>>> the issues surrounding the power supply pins. The +12 is the only one > >> that > >>>> is > >>>> sensitive enough to voltage (change in frequency vs voltage) to > >> contribute > >>>> to any > >>>> significant way to the overall stability. > >>>> > >>>> Bob > >>>> > >>>>> > >>>>> On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe < > >> scott.j.stobbe@gmail.com <javascript:;> > >>>> <javascript:;>> > >>>>> wrote: > >>>>> > >>>>>> Nick had mention that the -12V rail on the thunderbolt has the > poorest > >>>>>> PSRR with respect to frequency output, so I first took a look at the > >>>>>> venerable 7912. > >>>>>> > >>>>>> The first data-set was taken with a -13.5 VDC input. Attached is the > >> 0.1 > >>>>>> Hz to 10 Hz noise of an essentially quiescently loaded 7912, only a > >> 10k > >>>>>> resistor was added as load for preliminary evaluation. With a 60 dB > >>>> preamp > >>>>>> the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band > noise > >>>> is > >>>>>> 15 uVrms, which is about 1.3 ppm rms of the DC mean. > >>>>>> > >>>>>> In allan deviation terms, a quiescently loaded 7912 has a spot noise > >> of > >>>> 7 > >>>>>> uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580 ppb/rtHz. > >>>>>> Equivalently speaking, the flicker noise floor of an allan deviation > >>>> plot > >>>>>> would be sqrt(2*ln(2)) that figure to be 6.8E-7. > >>>>>> > >>>>>> Assuming a thunderbolt should be achieving 1/f floor of around > 1E-12, > >> it > >>>>>> would need a PSRR of at least 1 ppm/V. I'm sure someone has gone to > >> the > >>>>>> trouble of actually measuring it. > >>>>>> > >>>>>> So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't terrible > >>>>>> with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms, > >> only > >>>> one > >>>>>> order of magnitude off. > >>>>>> > >>>>>> The bad side of a 7912 is in long-term stability and tempCo, the > >> sample > >>>> I > >>>>>> tested had at least a 150 ppm/degC tempCo, which is going to put a > >>>> serious > >>>>>> lump/bump in the 10s tau to gps crossover point on an allan > deviation > >>>> plot. > >>>>>> > >>>>>> > >>>>>> > >>>>>> On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe < > >> scott.j.stobbe@gmail.com <javascript:;> > >>>> <javascript:;>> > >>>>>> wrote: > >>>>>> > >>>>>>> I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm > >>>>>>> interested to see the 1/f noise of a classic regulator, what load > >>>> current > >>>>>>> do you expect? I can bias a 7805 for the same load and measure the > >> 0.1 > >>>> to > >>>>>>> 10 Hz noise. > >>>>>>> > >>>>>>> Also if you have a digital scope without a very good builtin FFT, > >>>> octave > >>>>>>> would be one solution. > >>>>>>> > >>>>>>> On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts < > >>>>>>> time-nuts@febo.com <javascript:;> <javascript:;>> wrote: > >>>>>>> > >>>>>>>> Just an update. I’ve built the second prototype board (I skipped > >> over > >>>>>>>> the first design), and it’s powering my tbolt right now. > >>>>>>>> > >>>>>>>> The design calls for 15v in (though it would also work with > 13.8v). > >>>> The > >>>>>>>> +12 output comes from a D2PAK 7812. For +5, there is an AP1509 > buck > >>>>>>>> converter to make around 6.5 volts, then a DPAK 7805. For -12, > there > >>>> is an > >>>>>>>> MC34063 configured as an inverter to make around -13.75 volts and > >>>> then a > >>>>>>>> DPAK 7912. > >>>>>>>> > >>>>>>>> Steady-state, the system appears to be working just fine. The > >> AP1509’s > >>>>>>>> inductor and the D2PAK 7812 are just warm to the touch. > >>>>>>>> > >>>>>>>> I checked for noise and ripple on the outputs and it’s somewhere > >>>> around > >>>>>>>> ±2 mV or so generally. From what I can see on the scope, there’s > no > >>>> ripple > >>>>>>>> - it’s all high frequency noise. I am not absolutely certain that > >> the > >>>> noise > >>>>>>>> measurement represents real noise or the limits of my measuring > >>>> ability. > >>>>>>>> I’m just using the scope probes the scope came with, and 2 mV/div > is > >>>> its > >>>>>>>> lowest range. > >>>>>>>> > >>>>>>>> I haven’t compared the noise with the ex laptop supply that I was > >>>> using > >>>>>>>> before, but I’d have to believe it’s cleaner. I don’t really have > a > >>>> way to > >>>>>>>> check the oscillator’s before and after ADEV. My only other > >> reference > >>>> is an > >>>>>>>> FE5680A, and I think the thunderbolt’s going to be far better at > >>>> lower tau > >>>>>>>> (where this all matters). > >>>>>>>> > >>>>>>>> I know also that ±2 mV is still one and perhaps two orders of > >>>> magnitude > >>>>>>>> higher than some have called for. But before I attempt to reduce > the > >>>> noise > >>>>>>>> further, I’d like to know that there are real gains to be had. > Would > >>>>>>>> someone with a Thunderbolt and better output noise measuring > >>>> wherewithal be > >>>>>>>> willing to take a prototype and compare it with something that > does > >>>> have µV > >>>>>>>> levels of noise and ripple so I can get an idea of what there is > to > >>>> gain? > >>>>>>>> If you like, you can make such comparisons public - no secrets > here. > >>>>>>>> > >>>>>>>>> On Aug 30, 2016, at 10:37 PM, Nick Sayer <nsayer@kfu.com > >> <javascript:;> > >>>> <javascript:;>> wrote: > >>>>>>>>> > >>>>>>>>> > >>>>>>>>>> On Aug 30, 2016, at 8:48 PM, Cube Central < > cubecentral@gmail.com > >> <javascript:;> > >>>> <javascript:;>> > >>>>>>>> wrote: > >>>>>>>>>> > >>>>>>>>>> I would be interested, I think. Planning ahead for if the one I > >>>> have > >>>>>>>> for my Thunderbolt fails, I guess. Are there different models or > >>>> would a > >>>>>>>> photo of the input ports on mine be useful? > >>>>>>>>> > >>>>>>>>> Actually, what I had in mind is to just put a SIP4 header on the > >>>> board > >>>>>>>> for the output and people could wire the “last mile” themselves. > The > >>>> input > >>>>>>>> is a 2.1mm barrel connector. You use whatever 15W 12VDC wall wart > is > >>>> handy > >>>>>>>> and plug it right in. > >>>>>>>>> > >>>>>>>>> What it really amounts to is that you get +12 volts directly from > >> the > >>>>>>>> input, then there’s a buck converter to drop the +12 down to +5 > and > >> an > >>>>>>>> inverter to generate -12 from the +12. Those 3 voltages, plus a > >>>> ground go > >>>>>>>> to the SIP4. > >>>>>>>>> > >>>>>>>>> So it’s just two switching power supplies to turn a +12 volt only > >>>>>>>> supply into the three-way that the Thunderbolt wants. > >>>>>>>>> > >>>>>>>>> It’d be good for around 1500 mA @ 5V and around 50 mA @ -12 (the > >> +12 > >>>>>>>> spec is whatever is left from the source supply’s power spec) - > more > >>>> than > >>>>>>>> enough for a Thunderbolt. Probably enough for a hard disk or a > >>>> smallish PC. > >>>>>>>> _______________________________________________ > >>>>>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> > >> <javascript:;> > >>>>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/m > >>>>>>>> ailman/listinfo/time-nuts > >>>>>>>> and follow the instructions there. > >>>>>>>> > >>>>>>> > >>>>>>> > >>>>>> > >>>>> <7912_1PLC.png><7912_AllanDeviation.png><7912_ > >>>> TempCo.png>_______________________________________________ > >>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> > >> <javascript:;> > >>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ > >>>> mailman/listinfo/time-nuts > >>>>> and follow the instructions there. > >>>> > >>>> _______________________________________________ > >>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> > >> <javascript:;> > >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ > >>>> mailman/listinfo/time-nuts > >>>> and follow the instructions there. > >>>> > >>> _______________________________________________ > >>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> > >>> To unsubscribe, go to https://www.febo.com/cgi-bin/ > >> mailman/listinfo/time-nuts > >>> and follow the instructions there. > >> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@febo.com <javascript:;> > >> To unsubscribe, go to https://www.febo.com/cgi-bin/ > >> mailman/listinfo/time-nuts > >> and follow the instructions there. > >> > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > and follow the instructions there. >
BC
Bob Camp
Sat, Oct 22, 2016 5:41 PM

Hi

There is an enormous difference between regulation and it’s impact on
stability (which is what this drifted off into) and PSRR and it’s impact on
phase noise (which got lost at the tread moved on). PSRR does indeed
matter, but a fairly simple linear regulator (or pair of cheap ones or coil plus
pair) can take care of that.

The EFC reference in the TBolt comes from the TBolt board and not from the OCXO.
It is not in any way connected to the -12 V supply line.

Bob

On Oct 22, 2016, at 12:41 PM, Scott Stobbe scott.j.stobbe@gmail.com wrote:

Bob, this is good data and insights thanks for taking the time to share.
Ultimately the question Nick is hoping to answer is what is the point of
diminishing returns for voltage regulation. I think there are plenty of
folks on this list that have shared data suggesting a switching regulator
which meets these specifications:

Table 3-2 Power Specifications (Board-only)
+12v +/-10% max current 750 mA
+5v +/-5% max current 400 mA
-12v +/-10% max current 10 mA
Ripple
+5v 50 mV peak-to-peak, 15 mVrms
+/-12v 75 mV peak-to-peak, 20 mVrms

definitely has a degrading impact on the 10MHz phase noise.

I would happily agree that PSRR of an opamp buffering the EFC line is not
going to have a significant impact on performance with a jellybean LDO.

Do you know if the +- 5V reference is derived on board or is it from an EFC
reference provided by the ocxo?

On Sat, Oct 22, 2016 at 11:31 AM, Bob Camp kb8tq@n1k.org wrote:

Hi

The -12 V line is not the reference for the EFC. It is one supply into
the op amp circuit
that drives the EFC. Since the EFC is +/- 5V, there must be both a
positive and
a negative supply into the driver circuit for full output swing. The -12
can be anything between about
-13 and -7 without significantly impacting the function of the device. In
fact, people
with EFC voltages that are slightly positive have accidentally proven that
the
TBolt will work (lock up and function correctly) with zero volts on the
-12V line.

Bob

On Oct 22, 2016, at 10:24 AM, Scott Stobbe scott.j.stobbe@gmail.com

wrote:

It all depends what the -12V rail is for, some have said it directly
references the EFC dac. I would hope an ocxo would have a better tuning
gain on its efc pin than supply pin but maybe that's not always true.

On Saturday, 22 October 2016, Bob Camp kb8tq@n1k.org wrote:

Hi

The +12 goes straight to the OCXO. All OCXO’s have a voltage

sensitivity.

That sensitivity is much higher than voltage sensitivity is much higher
than
what you see on the other two supply pins.

Bob

On Oct 21, 2016, at 11:06 PM, Scott Stobbe <scott.j.stobbe@gmail.com

javascript:;> wrote:

Interesting, I would have thought that the +12V input would be

extremely

well regulated since its shared with the oven heater, I*R drops are

going

to show up every where, if your looking for uV levels of stability.
Just a connector has milliohms of contact resistance, let alone routing

and

cables...

On Friday, 21 October 2016, Bob Camp <kb8tq@n1k.org javascript:;>

wrote:

Hi

On Oct 21, 2016, at 7:44 PM, Scott Stobbe <scott.j.stobbe@gmail.com

javascript:;> wrote:

A little more data on the 7912.

The first plot shows the tempCo of the 7912 measured with ambient
temperature swings "7912_TempCo.png". Which is -150 ppm/degC.

The second plot is off a 7912 logged for an hour or so,

"7912_1PLC.png",

nothing too interesting here. However the environmental temperature

swing

of about 1 degC/hour is pretty conservative for a DUT sitting in free

air.

Finally, an allan devation plot looking at the normalized stability

of

a

7912 regulator "7912_AllanDeviation.png". Interestingly here, is, how

quick

a 15 mK/min temperature swing shoots above the 1/f floor, it's a

matter

of

seconds.

Now if your PSRR is 1 ppb/V or better, then all of this is

comfortably

below the intrinsic noise of a thunderbolt.

The main point is that the internal tempco of the TBolt it’s self is

much

larger than
the issues surrounding the power supply pins. The +12 is the only one

that

is
sensitive enough to voltage (change in frequency vs voltage) to

contribute

to any
significant way to the overall stability.

Bob

On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe <

wrote:

Nick had mention that the -12V rail on the thunderbolt has the

poorest

PSRR with respect to frequency output, so I first took a look at the
venerable 7912.

The first data-set was taken with a -13.5 VDC input. Attached is the

0.1

Hz to 10 Hz noise of an essentially quiescently loaded 7912, only a

10k

resistor was added as load for preliminary evaluation. With a 60 dB

preamp

the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band

noise

is

15 uVrms, which is about 1.3 ppm rms of the DC mean.

In allan deviation terms, a quiescently loaded 7912 has a spot noise

of

7

uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580 ppb/rtHz.
Equivalently speaking, the flicker noise floor of an allan deviation

plot

would be sqrt(2*ln(2)) that figure to be 6.8E-7.

Assuming a thunderbolt should be achieving 1/f floor of around

1E-12,

it

would need a PSRR of at least 1 ppm/V. I'm sure someone has gone to

the

trouble of actually measuring it.

So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't terrible
with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms,

only

one

order of magnitude off.

The bad side of a 7912 is in long-term stability and tempCo, the

sample

I

tested had at least a 150 ppm/degC tempCo, which is going to put a

serious

lump/bump in the 10s tau to gps crossover point on an allan

deviation

plot.

On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe <

wrote:

I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm
interested to see the 1/f noise of a classic regulator, what load

current

do you expect? I can bias a 7805 for the same load and measure the

0.1

to

10 Hz noise.

Also if you have a digital scope without a very good builtin FFT,

octave

would be one solution.

On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts <
time-nuts@febo.com javascript:; javascript:;> wrote:

Just an update. I’ve built the second prototype board (I skipped

over

the first design), and it’s powering my tbolt right now.

The design calls for 15v in (though it would also work with

13.8v).

The

+12 output comes from a D2PAK 7812. For +5, there is an AP1509

buck

converter to make around 6.5 volts, then a DPAK 7805. For -12,

there

is an

MC34063 configured as an inverter to make around -13.75 volts and

then a

DPAK 7912.

Steady-state, the system appears to be working just fine. The

AP1509’s

inductor and the D2PAK 7812 are just warm to the touch.

I checked for noise and ripple on the outputs and it’s somewhere

around

±2 mV or so generally. From what I can see on the scope, there’s

no

ripple

  • it’s all high frequency noise. I am not absolutely certain that

the

noise

measurement represents real noise or the limits of my measuring

ability.

I’m just using the scope probes the scope came with, and 2 mV/div

is

its

lowest range.

I haven’t compared the noise with the ex laptop supply that I was

using

before, but I’d have to believe it’s cleaner. I don’t really have

a

way to

check the oscillator’s before and after ADEV. My only other

reference

is an

FE5680A, and I think the thunderbolt’s going to be far better at

lower tau

(where this all matters).

I know also that ±2 mV is still one and perhaps two orders of

magnitude

higher than some have called for. But before I attempt to reduce

the

noise

further, I’d like to know that there are real gains to be had.

Would

someone with a Thunderbolt and better output noise measuring

wherewithal be

willing to take a prototype and compare it with something that

does

have µV

levels of noise and ripple so I can get an idea of what there is

to

gain?

If you like, you can make such comparisons public - no secrets

here.

On Aug 30, 2016, at 10:37 PM, Nick Sayer <nsayer@kfu.com

javascript:;> wrote:

On Aug 30, 2016, at 8:48 PM, Cube Central <

wrote:

I would be interested, I think.  Planning ahead for if the one I

have

for my Thunderbolt fails, I guess.  Are there different models or

would a

photo of the input ports on mine be useful?

Actually, what I had in mind is to just put a SIP4 header on the

board

for the output and people could wire the “last mile” themselves.

The

input

is a 2.1mm barrel connector. You use whatever 15W 12VDC wall wart

is

handy

and plug it right in.

What it really amounts to is that you get +12 volts directly from

the

input, then there’s a buck converter to drop the +12 down to +5

and

an

inverter to generate -12 from the +12. Those 3 voltages, plus a

ground go

to the SIP4.

So it’s just two switching power supplies to turn a +12 volt only

supply into the three-way that the Thunderbolt wants.

It’d be good for around 1500 mA @ 5V and around 50 mA @ -12 (the

+12

spec is whatever is left from the source supply’s power spec) -

more

than

enough for a Thunderbolt. Probably enough for a hard disk or a

smallish PC.


time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

<7912_1PLC.png><7912_AllanDeviation.png><7912_

TempCo.png>_______________________________________________

time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi There is an *enormous* difference between regulation and it’s impact on stability (which is what this drifted off into) and PSRR and it’s impact on phase noise (which got lost at the tread moved on). PSRR does indeed matter, but a fairly simple linear regulator (or pair of cheap ones or coil plus pair) can take care of that. The EFC reference in the TBolt comes from the TBolt board and not from the OCXO. It is not in any way connected to the -12 V supply line. Bob > On Oct 22, 2016, at 12:41 PM, Scott Stobbe <scott.j.stobbe@gmail.com> wrote: > > Bob, this is good data and insights thanks for taking the time to share. > Ultimately the question Nick is hoping to answer is what is the point of > diminishing returns for voltage regulation. I think there are plenty of > folks on this list that have shared data suggesting a switching regulator > which meets these specifications: > > Table 3-2 Power Specifications (Board-only) > +12v +/-10% max current 750 mA > +5v +/-5% max current 400 mA > -12v +/-10% max current 10 mA > Ripple > +5v 50 mV peak-to-peak, 15 mVrms > +/-12v 75 mV peak-to-peak, 20 mVrms > > definitely has a degrading impact on the 10MHz phase noise. > > I would happily agree that PSRR of an opamp buffering the EFC line is not > going to have a significant impact on performance with a jellybean LDO. > > Do you know if the +- 5V reference is derived on board or is it from an EFC > reference provided by the ocxo? > > On Sat, Oct 22, 2016 at 11:31 AM, Bob Camp <kb8tq@n1k.org> wrote: > >> Hi >> >> The -12 V line is not the reference for the EFC. It *is* one supply into >> the op amp circuit >> that drives the EFC. Since the EFC is +/- 5V, there must be both a >> positive and >> a negative supply into the driver circuit for full output swing. The -12 >> can be anything between about >> -13 and -7 without significantly impacting the function of the device. In >> fact, people >> with EFC voltages that are slightly positive have accidentally proven that >> the >> TBolt will work (lock up and function correctly) with zero volts on the >> -12V line. >> >> Bob >> >>> On Oct 22, 2016, at 10:24 AM, Scott Stobbe <scott.j.stobbe@gmail.com> >> wrote: >>> >>> It all depends what the -12V rail is for, some have said it directly >>> references the EFC dac. I would hope an ocxo would have a better tuning >>> gain on its efc pin than supply pin but maybe that's not always true. >>> >>> On Saturday, 22 October 2016, Bob Camp <kb8tq@n1k.org> wrote: >>> >>>> Hi >>>> >>>> The +12 goes straight to the OCXO. All OCXO’s have a voltage >> sensitivity. >>>> That sensitivity is much higher than voltage sensitivity is much higher >>>> than >>>> what you see on the other two supply pins. >>>> >>>> Bob >>>> >>>>> On Oct 21, 2016, at 11:06 PM, Scott Stobbe <scott.j.stobbe@gmail.com >>>> <javascript:;>> wrote: >>>>> >>>>> Interesting, I would have thought that the +12V input would be >> extremely >>>>> well regulated since its shared with the oven heater, I*R drops are >> going >>>>> to show up every where, if your looking for uV levels of stability. >>>>> Just a connector has milliohms of contact resistance, let alone routing >>>> and >>>>> cables... >>>>> >>>>> On Friday, 21 October 2016, Bob Camp <kb8tq@n1k.org <javascript:;>> >>>> wrote: >>>>> >>>>>> Hi >>>>>> >>>>>>> On Oct 21, 2016, at 7:44 PM, Scott Stobbe <scott.j.stobbe@gmail.com >>>> <javascript:;> >>>>>> <javascript:;>> wrote: >>>>>>> >>>>>>> A little more data on the 7912. >>>>>>> >>>>>>> The first plot shows the tempCo of the 7912 measured with ambient >>>>>>> temperature swings "7912_TempCo.png". Which is -150 ppm/degC. >>>>>>> >>>>>>> The second plot is off a 7912 logged for an hour or so, >>>> "7912_1PLC.png", >>>>>>> nothing too interesting here. However the environmental temperature >>>> swing >>>>>>> of about 1 degC/hour is pretty conservative for a DUT sitting in free >>>>>> air. >>>>>>> >>>>>>> Finally, an allan devation plot looking at the normalized stability >> of >>>> a >>>>>>> 7912 regulator "7912_AllanDeviation.png". Interestingly here, is, how >>>>>> quick >>>>>>> a 15 mK/min temperature swing shoots above the 1/f floor, it's a >> matter >>>>>> of >>>>>>> seconds. >>>>>>> >>>>>>> Now if your PSRR is 1 ppb/V or better, then all of this is >> comfortably >>>>>>> below the intrinsic noise of a thunderbolt. >>>>>> >>>>>> The main point is that the internal tempco of the TBolt it’s self is >>>> much >>>>>> larger than >>>>>> the issues surrounding the power supply pins. The +12 is the only one >>>> that >>>>>> is >>>>>> sensitive enough to voltage (change in frequency vs voltage) to >>>> contribute >>>>>> to any >>>>>> significant way to the overall stability. >>>>>> >>>>>> Bob >>>>>> >>>>>>> >>>>>>> On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe < >>>> scott.j.stobbe@gmail.com <javascript:;> >>>>>> <javascript:;>> >>>>>>> wrote: >>>>>>> >>>>>>>> Nick had mention that the -12V rail on the thunderbolt has the >> poorest >>>>>>>> PSRR with respect to frequency output, so I first took a look at the >>>>>>>> venerable 7912. >>>>>>>> >>>>>>>> The first data-set was taken with a -13.5 VDC input. Attached is the >>>> 0.1 >>>>>>>> Hz to 10 Hz noise of an essentially quiescently loaded 7912, only a >>>> 10k >>>>>>>> resistor was added as load for preliminary evaluation. With a 60 dB >>>>>> preamp >>>>>>>> the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band >> noise >>>>>> is >>>>>>>> 15 uVrms, which is about 1.3 ppm rms of the DC mean. >>>>>>>> >>>>>>>> In allan deviation terms, a quiescently loaded 7912 has a spot noise >>>> of >>>>>> 7 >>>>>>>> uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580 ppb/rtHz. >>>>>>>> Equivalently speaking, the flicker noise floor of an allan deviation >>>>>> plot >>>>>>>> would be sqrt(2*ln(2)) that figure to be 6.8E-7. >>>>>>>> >>>>>>>> Assuming a thunderbolt should be achieving 1/f floor of around >> 1E-12, >>>> it >>>>>>>> would need a PSRR of at least 1 ppm/V. I'm sure someone has gone to >>>> the >>>>>>>> trouble of actually measuring it. >>>>>>>> >>>>>>>> So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't terrible >>>>>>>> with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms, >>>> only >>>>>> one >>>>>>>> order of magnitude off. >>>>>>>> >>>>>>>> The bad side of a 7912 is in long-term stability and tempCo, the >>>> sample >>>>>> I >>>>>>>> tested had at least a 150 ppm/degC tempCo, which is going to put a >>>>>> serious >>>>>>>> lump/bump in the 10s tau to gps crossover point on an allan >> deviation >>>>>> plot. >>>>>>>> >>>>>>>> >>>>>>>> >>>>>>>> On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe < >>>> scott.j.stobbe@gmail.com <javascript:;> >>>>>> <javascript:;>> >>>>>>>> wrote: >>>>>>>> >>>>>>>>> I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm >>>>>>>>> interested to see the 1/f noise of a classic regulator, what load >>>>>> current >>>>>>>>> do you expect? I can bias a 7805 for the same load and measure the >>>> 0.1 >>>>>> to >>>>>>>>> 10 Hz noise. >>>>>>>>> >>>>>>>>> Also if you have a digital scope without a very good builtin FFT, >>>>>> octave >>>>>>>>> would be one solution. >>>>>>>>> >>>>>>>>> On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts < >>>>>>>>> time-nuts@febo.com <javascript:;> <javascript:;>> wrote: >>>>>>>>> >>>>>>>>>> Just an update. I’ve built the second prototype board (I skipped >>>> over >>>>>>>>>> the first design), and it’s powering my tbolt right now. >>>>>>>>>> >>>>>>>>>> The design calls for 15v in (though it would also work with >> 13.8v). >>>>>> The >>>>>>>>>> +12 output comes from a D2PAK 7812. For +5, there is an AP1509 >> buck >>>>>>>>>> converter to make around 6.5 volts, then a DPAK 7805. For -12, >> there >>>>>> is an >>>>>>>>>> MC34063 configured as an inverter to make around -13.75 volts and >>>>>> then a >>>>>>>>>> DPAK 7912. >>>>>>>>>> >>>>>>>>>> Steady-state, the system appears to be working just fine. The >>>> AP1509’s >>>>>>>>>> inductor and the D2PAK 7812 are just warm to the touch. >>>>>>>>>> >>>>>>>>>> I checked for noise and ripple on the outputs and it’s somewhere >>>>>> around >>>>>>>>>> ±2 mV or so generally. From what I can see on the scope, there’s >> no >>>>>> ripple >>>>>>>>>> - it’s all high frequency noise. I am not absolutely certain that >>>> the >>>>>> noise >>>>>>>>>> measurement represents real noise or the limits of my measuring >>>>>> ability. >>>>>>>>>> I’m just using the scope probes the scope came with, and 2 mV/div >> is >>>>>> its >>>>>>>>>> lowest range. >>>>>>>>>> >>>>>>>>>> I haven’t compared the noise with the ex laptop supply that I was >>>>>> using >>>>>>>>>> before, but I’d have to believe it’s cleaner. I don’t really have >> a >>>>>> way to >>>>>>>>>> check the oscillator’s before and after ADEV. My only other >>>> reference >>>>>> is an >>>>>>>>>> FE5680A, and I think the thunderbolt’s going to be far better at >>>>>> lower tau >>>>>>>>>> (where this all matters). >>>>>>>>>> >>>>>>>>>> I know also that ±2 mV is still one and perhaps two orders of >>>>>> magnitude >>>>>>>>>> higher than some have called for. But before I attempt to reduce >> the >>>>>> noise >>>>>>>>>> further, I’d like to know that there are real gains to be had. >> Would >>>>>>>>>> someone with a Thunderbolt and better output noise measuring >>>>>> wherewithal be >>>>>>>>>> willing to take a prototype and compare it with something that >> does >>>>>> have µV >>>>>>>>>> levels of noise and ripple so I can get an idea of what there is >> to >>>>>> gain? >>>>>>>>>> If you like, you can make such comparisons public - no secrets >> here. >>>>>>>>>> >>>>>>>>>>> On Aug 30, 2016, at 10:37 PM, Nick Sayer <nsayer@kfu.com >>>> <javascript:;> >>>>>> <javascript:;>> wrote: >>>>>>>>>>> >>>>>>>>>>> >>>>>>>>>>>> On Aug 30, 2016, at 8:48 PM, Cube Central < >> cubecentral@gmail.com >>>> <javascript:;> >>>>>> <javascript:;>> >>>>>>>>>> wrote: >>>>>>>>>>>> >>>>>>>>>>>> I would be interested, I think. Planning ahead for if the one I >>>>>> have >>>>>>>>>> for my Thunderbolt fails, I guess. Are there different models or >>>>>> would a >>>>>>>>>> photo of the input ports on mine be useful? >>>>>>>>>>> >>>>>>>>>>> Actually, what I had in mind is to just put a SIP4 header on the >>>>>> board >>>>>>>>>> for the output and people could wire the “last mile” themselves. >> The >>>>>> input >>>>>>>>>> is a 2.1mm barrel connector. You use whatever 15W 12VDC wall wart >> is >>>>>> handy >>>>>>>>>> and plug it right in. >>>>>>>>>>> >>>>>>>>>>> What it really amounts to is that you get +12 volts directly from >>>> the >>>>>>>>>> input, then there’s a buck converter to drop the +12 down to +5 >> and >>>> an >>>>>>>>>> inverter to generate -12 from the +12. Those 3 voltages, plus a >>>>>> ground go >>>>>>>>>> to the SIP4. >>>>>>>>>>> >>>>>>>>>>> So it’s just two switching power supplies to turn a +12 volt only >>>>>>>>>> supply into the three-way that the Thunderbolt wants. >>>>>>>>>>> >>>>>>>>>>> It’d be good for around 1500 mA @ 5V and around 50 mA @ -12 (the >>>> +12 >>>>>>>>>> spec is whatever is left from the source supply’s power spec) - >> more >>>>>> than >>>>>>>>>> enough for a Thunderbolt. Probably enough for a hard disk or a >>>>>> smallish PC. >>>>>>>>>> _______________________________________________ >>>>>>>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >>>> <javascript:;> >>>>>>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/m >>>>>>>>>> ailman/listinfo/time-nuts >>>>>>>>>> and follow the instructions there. >>>>>>>>>> >>>>>>>>> >>>>>>>>> >>>>>>>> >>>>>>> <7912_1PLC.png><7912_AllanDeviation.png><7912_ >>>>>> TempCo.png>_______________________________________________ >>>>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >>>> <javascript:;> >>>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>>>>> mailman/listinfo/time-nuts >>>>>>> and follow the instructions there. >>>>>> >>>>>> _______________________________________________ >>>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >>>> <javascript:;> >>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>>>>> mailman/listinfo/time-nuts >>>>>> and follow the instructions there. >>>>>> >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>>> mailman/listinfo/time-nuts >>>>> and follow the instructions there. >>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>>> mailman/listinfo/time-nuts >>>> and follow the instructions there. >>>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >> mailman/listinfo/time-nuts >>> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/ >> mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
SS
Scott Stobbe
Sat, Oct 22, 2016 8:15 PM

I think we are coming up to the noise floor here. We have identified a
potential signal A from a 7912 (likely similar to a 7812, no guarentee).
The remaining question is if you apply signal A to the power pin of a
thunderbolt what is signal B added to the output. We know where we want
signal B to be, below the noise. Knowing the PSSR for each of the power
inputs would be one way to answer this question.

On Saturday, 22 October 2016, Bob Camp kb8tq@n1k.org wrote:

Hi

There is an enormous difference between regulation and it’s impact on
stability (which is what this drifted off into) and PSRR and it’s impact on
phase noise (which got lost at the tread moved on). PSRR does indeed
matter, but a fairly simple linear regulator (or pair of cheap ones or
coil plus
pair) can take care of that.

The EFC reference in the TBolt comes from the TBolt board and not from the
OCXO.
It is not in any way connected to the -12 V supply line.

Bob

On Oct 22, 2016, at 12:41 PM, Scott Stobbe <scott.j.stobbe@gmail.com

javascript:;> wrote:

Bob, this is good data and insights thanks for taking the time to share.
Ultimately the question Nick is hoping to answer is what is the point of
diminishing returns for voltage regulation. I think there are plenty of
folks on this list that have shared data suggesting a switching regulator
which meets these specifications:

Table 3-2 Power Specifications (Board-only)
+12v +/-10% max current 750 mA
+5v +/-5% max current 400 mA
-12v +/-10% max current 10 mA
Ripple
+5v 50 mV peak-to-peak, 15 mVrms
+/-12v 75 mV peak-to-peak, 20 mVrms

definitely has a degrading impact on the 10MHz phase noise.

I would happily agree that PSRR of an opamp buffering the EFC line is not
going to have a significant impact on performance with a jellybean LDO.

Do you know if the +- 5V reference is derived on board or is it from an

EFC

reference provided by the ocxo?

On Sat, Oct 22, 2016 at 11:31 AM, Bob Camp <kb8tq@n1k.org javascript:;>

wrote:

Hi

The -12 V line is not the reference for the EFC. It is one supply into
the op amp circuit
that drives the EFC. Since the EFC is +/- 5V, there must be both a
positive and
a negative supply into the driver circuit for full output swing. The -12
can be anything between about
-13 and -7 without significantly impacting the function of the device.

In

fact, people
with EFC voltages that are slightly positive have accidentally proven

that

the
TBolt will work (lock up and function correctly) with zero volts on the
-12V line.

Bob

On Oct 22, 2016, at 10:24 AM, Scott Stobbe <scott.j.stobbe@gmail.com

wrote:

It all depends what the -12V rail is for, some have said it directly
references the EFC dac. I would hope an ocxo would have a better tuning
gain on its efc pin than supply pin but maybe that's not always true.

On Saturday, 22 October 2016, Bob Camp <kb8tq@n1k.org javascript:;>

wrote:

Hi

The +12 goes straight to the OCXO. All OCXO’s have a voltage

sensitivity.

That sensitivity is much higher than voltage sensitivity is much

higher

than
what you see on the other two supply pins.

Bob

On Oct 21, 2016, at 11:06 PM, Scott Stobbe <scott.j.stobbe@gmail.com

javascript:;> wrote:

Interesting, I would have thought that the +12V input would be

extremely

well regulated since its shared with the oven heater, I*R drops are

going

to show up every where, if your looking for uV levels of stability.
Just a connector has milliohms of contact resistance, let alone

routing

and

cables...

On Friday, 21 October 2016, Bob Camp <kb8tq@n1k.org javascript:;

wrote:

Hi

On Oct 21, 2016, at 7:44 PM, Scott Stobbe <

javascript:;> wrote:

A little more data on the 7912.

The first plot shows the tempCo of the 7912 measured with ambient
temperature swings "7912_TempCo.png". Which is -150 ppm/degC.

The second plot is off a 7912 logged for an hour or so,

"7912_1PLC.png",

nothing too interesting here. However the environmental temperature

swing

of about 1 degC/hour is pretty conservative for a DUT sitting in

free

air.

Finally, an allan devation plot looking at the normalized stability

of

a

7912 regulator "7912_AllanDeviation.png". Interestingly here, is,

how

quick

a 15 mK/min temperature swing shoots above the 1/f floor, it's a

matter

of

seconds.

Now if your PSRR is 1 ppb/V or better, then all of this is

comfortably

below the intrinsic noise of a thunderbolt.

The main point is that the internal tempco of the TBolt it’s self is

much

larger than
the issues surrounding the power supply pins. The +12 is the only

one

that

is
sensitive enough to voltage (change in frequency vs voltage) to

contribute

to any
significant way to the overall stability.

Bob

On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe <

wrote:

Nick had mention that the -12V rail on the thunderbolt has the

poorest

PSRR with respect to frequency output, so I first took a look at

the

venerable 7912.

The first data-set was taken with a -13.5 VDC input. Attached is

the

0.1

Hz to 10 Hz noise of an essentially quiescently loaded 7912, only

a

10k

resistor was added as load for preliminary evaluation. With a 60

dB

preamp

the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band

noise

is

15 uVrms, which is about 1.3 ppm rms of the DC mean.

In allan deviation terms, a quiescently loaded 7912 has a spot

noise

of

7

uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580

ppb/rtHz.

Equivalently speaking, the flicker noise floor of an allan

deviation

plot

would be sqrt(2*ln(2)) that figure to be 6.8E-7.

Assuming a thunderbolt should be achieving 1/f floor of around

1E-12,

it

would need a PSRR of at least 1 ppm/V. I'm sure someone has gone

to

the

trouble of actually measuring it.

So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't

terrible

with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms,

only

one

order of magnitude off.

The bad side of a 7912 is in long-term stability and tempCo, the

sample

I

tested had at least a 150 ppm/degC tempCo, which is going to put a

serious

lump/bump in the 10s tau to gps crossover point on an allan

deviation

plot.

On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe <

wrote:

I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm
interested to see the 1/f noise of a classic regulator, what load

current

do you expect? I can bias a 7805 for the same load and measure

the

0.1

to

10 Hz noise.

Also if you have a digital scope without a very good builtin FFT,

octave

would be one solution.

On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts <
time-nuts@febo.com javascript:; javascript:;

javascript:;> wrote:

Just an update. I’ve built the second prototype board (I skipped

over

the first design), and it’s powering my tbolt right now.

The design calls for 15v in (though it would also work with

13.8v).

The

+12 output comes from a D2PAK 7812. For +5, there is an AP1509

buck

converter to make around 6.5 volts, then a DPAK 7805. For -12,

there

is an

MC34063 configured as an inverter to make around -13.75 volts

and

then a

DPAK 7912.

Steady-state, the system appears to be working just fine. The

AP1509’s

inductor and the D2PAK 7812 are just warm to the touch.

I checked for noise and ripple on the outputs and it’s somewhere

around

±2 mV or so generally. From what I can see on the scope, there’s

no

ripple

  • it’s all high frequency noise. I am not absolutely certain

that

the

noise

measurement represents real noise or the limits of my measuring

ability.

I’m just using the scope probes the scope came with, and 2

mV/div

is

its

lowest range.

I haven’t compared the noise with the ex laptop supply that I

was

using

before, but I’d have to believe it’s cleaner. I don’t really

have

a

way to

check the oscillator’s before and after ADEV. My only other

reference

is an

FE5680A, and I think the thunderbolt’s going to be far better at

lower tau

(where this all matters).

I know also that ±2 mV is still one and perhaps two orders of

magnitude

higher than some have called for. But before I attempt to reduce

the

noise

further, I’d like to know that there are real gains to be had.

Would

someone with a Thunderbolt and better output noise measuring

wherewithal be

willing to take a prototype and compare it with something that

does

have µV

levels of noise and ripple so I can get an idea of what there is

to

gain?

If you like, you can make such comparisons public - no secrets

here.

On Aug 30, 2016, at 10:37 PM, Nick Sayer <nsayer@kfu.com

javascript:;> wrote:

On Aug 30, 2016, at 8:48 PM, Cube Central <

wrote:

I would be interested, I think.  Planning ahead for if the

one I

have

for my Thunderbolt fails, I guess.  Are there different models

or

would a

photo of the input ports on mine be useful?

Actually, what I had in mind is to just put a SIP4 header on

the

board

for the output and people could wire the “last mile” themselves.

The

input

is a 2.1mm barrel connector. You use whatever 15W 12VDC wall

wart

is

handy

and plug it right in.

What it really amounts to is that you get +12 volts directly

from

the

input, then there’s a buck converter to drop the +12 down to +5

and

an

inverter to generate -12 from the +12. Those 3 voltages, plus a

ground go

to the SIP4.

So it’s just two switching power supplies to turn a +12 volt

only

supply into the three-way that the Thunderbolt wants.

It’d be good for around 1500 mA @ 5V and around 50 mA @ -12

(the

+12

spec is whatever is left from the source supply’s power spec) -

more

than

enough for a Thunderbolt. Probably enough for a hard disk or a

smallish PC.


time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

<7912_1PLC.png><7912_AllanDeviation.png><7912_

TempCo.png>_______________________________________________

time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

I think we are coming up to the noise floor here. We have identified a potential signal A from a 7912 (likely similar to a 7812, no guarentee). The remaining question is if you apply signal A to the power pin of a thunderbolt what is signal B added to the output. We know where we want signal B to be, below the noise. Knowing the PSSR for each of the power inputs would be one way to answer this question. On Saturday, 22 October 2016, Bob Camp <kb8tq@n1k.org> wrote: > Hi > > There is an *enormous* difference between regulation and it’s impact on > stability (which is what this drifted off into) and PSRR and it’s impact on > phase noise (which got lost at the tread moved on). PSRR does indeed > matter, but a fairly simple linear regulator (or pair of cheap ones or > coil plus > pair) can take care of that. > > The EFC reference in the TBolt comes from the TBolt board and not from the > OCXO. > It is not in any way connected to the -12 V supply line. > > Bob > > > > > On Oct 22, 2016, at 12:41 PM, Scott Stobbe <scott.j.stobbe@gmail.com > <javascript:;>> wrote: > > > > Bob, this is good data and insights thanks for taking the time to share. > > Ultimately the question Nick is hoping to answer is what is the point of > > diminishing returns for voltage regulation. I think there are plenty of > > folks on this list that have shared data suggesting a switching regulator > > which meets these specifications: > > > > Table 3-2 Power Specifications (Board-only) > > +12v +/-10% max current 750 mA > > +5v +/-5% max current 400 mA > > -12v +/-10% max current 10 mA > > Ripple > > +5v 50 mV peak-to-peak, 15 mVrms > > +/-12v 75 mV peak-to-peak, 20 mVrms > > > > definitely has a degrading impact on the 10MHz phase noise. > > > > I would happily agree that PSRR of an opamp buffering the EFC line is not > > going to have a significant impact on performance with a jellybean LDO. > > > > Do you know if the +- 5V reference is derived on board or is it from an > EFC > > reference provided by the ocxo? > > > > On Sat, Oct 22, 2016 at 11:31 AM, Bob Camp <kb8tq@n1k.org <javascript:;>> > wrote: > > > >> Hi > >> > >> The -12 V line is not the reference for the EFC. It *is* one supply into > >> the op amp circuit > >> that drives the EFC. Since the EFC is +/- 5V, there must be both a > >> positive and > >> a negative supply into the driver circuit for full output swing. The -12 > >> can be anything between about > >> -13 and -7 without significantly impacting the function of the device. > In > >> fact, people > >> with EFC voltages that are slightly positive have accidentally proven > that > >> the > >> TBolt will work (lock up and function correctly) with zero volts on the > >> -12V line. > >> > >> Bob > >> > >>> On Oct 22, 2016, at 10:24 AM, Scott Stobbe <scott.j.stobbe@gmail.com > <javascript:;>> > >> wrote: > >>> > >>> It all depends what the -12V rail is for, some have said it directly > >>> references the EFC dac. I would hope an ocxo would have a better tuning > >>> gain on its efc pin than supply pin but maybe that's not always true. > >>> > >>> On Saturday, 22 October 2016, Bob Camp <kb8tq@n1k.org <javascript:;>> > wrote: > >>> > >>>> Hi > >>>> > >>>> The +12 goes straight to the OCXO. All OCXO’s have a voltage > >> sensitivity. > >>>> That sensitivity is much higher than voltage sensitivity is much > higher > >>>> than > >>>> what you see on the other two supply pins. > >>>> > >>>> Bob > >>>> > >>>>> On Oct 21, 2016, at 11:06 PM, Scott Stobbe <scott.j.stobbe@gmail.com > <javascript:;> > >>>> <javascript:;>> wrote: > >>>>> > >>>>> Interesting, I would have thought that the +12V input would be > >> extremely > >>>>> well regulated since its shared with the oven heater, I*R drops are > >> going > >>>>> to show up every where, if your looking for uV levels of stability. > >>>>> Just a connector has milliohms of contact resistance, let alone > routing > >>>> and > >>>>> cables... > >>>>> > >>>>> On Friday, 21 October 2016, Bob Camp <kb8tq@n1k.org <javascript:;> > <javascript:;>> > >>>> wrote: > >>>>> > >>>>>> Hi > >>>>>> > >>>>>>> On Oct 21, 2016, at 7:44 PM, Scott Stobbe < > scott.j.stobbe@gmail.com <javascript:;> > >>>> <javascript:;> > >>>>>> <javascript:;>> wrote: > >>>>>>> > >>>>>>> A little more data on the 7912. > >>>>>>> > >>>>>>> The first plot shows the tempCo of the 7912 measured with ambient > >>>>>>> temperature swings "7912_TempCo.png". Which is -150 ppm/degC. > >>>>>>> > >>>>>>> The second plot is off a 7912 logged for an hour or so, > >>>> "7912_1PLC.png", > >>>>>>> nothing too interesting here. However the environmental temperature > >>>> swing > >>>>>>> of about 1 degC/hour is pretty conservative for a DUT sitting in > free > >>>>>> air. > >>>>>>> > >>>>>>> Finally, an allan devation plot looking at the normalized stability > >> of > >>>> a > >>>>>>> 7912 regulator "7912_AllanDeviation.png". Interestingly here, is, > how > >>>>>> quick > >>>>>>> a 15 mK/min temperature swing shoots above the 1/f floor, it's a > >> matter > >>>>>> of > >>>>>>> seconds. > >>>>>>> > >>>>>>> Now if your PSRR is 1 ppb/V or better, then all of this is > >> comfortably > >>>>>>> below the intrinsic noise of a thunderbolt. > >>>>>> > >>>>>> The main point is that the internal tempco of the TBolt it’s self is > >>>> much > >>>>>> larger than > >>>>>> the issues surrounding the power supply pins. The +12 is the only > one > >>>> that > >>>>>> is > >>>>>> sensitive enough to voltage (change in frequency vs voltage) to > >>>> contribute > >>>>>> to any > >>>>>> significant way to the overall stability. > >>>>>> > >>>>>> Bob > >>>>>> > >>>>>>> > >>>>>>> On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe < > >>>> scott.j.stobbe@gmail.com <javascript:;> <javascript:;> > >>>>>> <javascript:;>> > >>>>>>> wrote: > >>>>>>> > >>>>>>>> Nick had mention that the -12V rail on the thunderbolt has the > >> poorest > >>>>>>>> PSRR with respect to frequency output, so I first took a look at > the > >>>>>>>> venerable 7912. > >>>>>>>> > >>>>>>>> The first data-set was taken with a -13.5 VDC input. Attached is > the > >>>> 0.1 > >>>>>>>> Hz to 10 Hz noise of an essentially quiescently loaded 7912, only > a > >>>> 10k > >>>>>>>> resistor was added as load for preliminary evaluation. With a 60 > dB > >>>>>> preamp > >>>>>>>> the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band > >> noise > >>>>>> is > >>>>>>>> 15 uVrms, which is about 1.3 ppm rms of the DC mean. > >>>>>>>> > >>>>>>>> In allan deviation terms, a quiescently loaded 7912 has a spot > noise > >>>> of > >>>>>> 7 > >>>>>>>> uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580 > ppb/rtHz. > >>>>>>>> Equivalently speaking, the flicker noise floor of an allan > deviation > >>>>>> plot > >>>>>>>> would be sqrt(2*ln(2)) that figure to be 6.8E-7. > >>>>>>>> > >>>>>>>> Assuming a thunderbolt should be achieving 1/f floor of around > >> 1E-12, > >>>> it > >>>>>>>> would need a PSRR of at least 1 ppm/V. I'm sure someone has gone > to > >>>> the > >>>>>>>> trouble of actually measuring it. > >>>>>>>> > >>>>>>>> So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't > terrible > >>>>>>>> with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms, > >>>> only > >>>>>> one > >>>>>>>> order of magnitude off. > >>>>>>>> > >>>>>>>> The bad side of a 7912 is in long-term stability and tempCo, the > >>>> sample > >>>>>> I > >>>>>>>> tested had at least a 150 ppm/degC tempCo, which is going to put a > >>>>>> serious > >>>>>>>> lump/bump in the 10s tau to gps crossover point on an allan > >> deviation > >>>>>> plot. > >>>>>>>> > >>>>>>>> > >>>>>>>> > >>>>>>>> On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe < > >>>> scott.j.stobbe@gmail.com <javascript:;> <javascript:;> > >>>>>> <javascript:;>> > >>>>>>>> wrote: > >>>>>>>> > >>>>>>>>> I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm > >>>>>>>>> interested to see the 1/f noise of a classic regulator, what load > >>>>>> current > >>>>>>>>> do you expect? I can bias a 7805 for the same load and measure > the > >>>> 0.1 > >>>>>> to > >>>>>>>>> 10 Hz noise. > >>>>>>>>> > >>>>>>>>> Also if you have a digital scope without a very good builtin FFT, > >>>>>> octave > >>>>>>>>> would be one solution. > >>>>>>>>> > >>>>>>>>> On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts < > >>>>>>>>> time-nuts@febo.com <javascript:;> <javascript:;> > <javascript:;>> wrote: > >>>>>>>>> > >>>>>>>>>> Just an update. I’ve built the second prototype board (I skipped > >>>> over > >>>>>>>>>> the first design), and it’s powering my tbolt right now. > >>>>>>>>>> > >>>>>>>>>> The design calls for 15v in (though it would also work with > >> 13.8v). > >>>>>> The > >>>>>>>>>> +12 output comes from a D2PAK 7812. For +5, there is an AP1509 > >> buck > >>>>>>>>>> converter to make around 6.5 volts, then a DPAK 7805. For -12, > >> there > >>>>>> is an > >>>>>>>>>> MC34063 configured as an inverter to make around -13.75 volts > and > >>>>>> then a > >>>>>>>>>> DPAK 7912. > >>>>>>>>>> > >>>>>>>>>> Steady-state, the system appears to be working just fine. The > >>>> AP1509’s > >>>>>>>>>> inductor and the D2PAK 7812 are just warm to the touch. > >>>>>>>>>> > >>>>>>>>>> I checked for noise and ripple on the outputs and it’s somewhere > >>>>>> around > >>>>>>>>>> ±2 mV or so generally. From what I can see on the scope, there’s > >> no > >>>>>> ripple > >>>>>>>>>> - it’s all high frequency noise. I am not absolutely certain > that > >>>> the > >>>>>> noise > >>>>>>>>>> measurement represents real noise or the limits of my measuring > >>>>>> ability. > >>>>>>>>>> I’m just using the scope probes the scope came with, and 2 > mV/div > >> is > >>>>>> its > >>>>>>>>>> lowest range. > >>>>>>>>>> > >>>>>>>>>> I haven’t compared the noise with the ex laptop supply that I > was > >>>>>> using > >>>>>>>>>> before, but I’d have to believe it’s cleaner. I don’t really > have > >> a > >>>>>> way to > >>>>>>>>>> check the oscillator’s before and after ADEV. My only other > >>>> reference > >>>>>> is an > >>>>>>>>>> FE5680A, and I think the thunderbolt’s going to be far better at > >>>>>> lower tau > >>>>>>>>>> (where this all matters). > >>>>>>>>>> > >>>>>>>>>> I know also that ±2 mV is still one and perhaps two orders of > >>>>>> magnitude > >>>>>>>>>> higher than some have called for. But before I attempt to reduce > >> the > >>>>>> noise > >>>>>>>>>> further, I’d like to know that there are real gains to be had. > >> Would > >>>>>>>>>> someone with a Thunderbolt and better output noise measuring > >>>>>> wherewithal be > >>>>>>>>>> willing to take a prototype and compare it with something that > >> does > >>>>>> have µV > >>>>>>>>>> levels of noise and ripple so I can get an idea of what there is > >> to > >>>>>> gain? > >>>>>>>>>> If you like, you can make such comparisons public - no secrets > >> here. > >>>>>>>>>> > >>>>>>>>>>> On Aug 30, 2016, at 10:37 PM, Nick Sayer <nsayer@kfu.com > <javascript:;> > >>>> <javascript:;> > >>>>>> <javascript:;>> wrote: > >>>>>>>>>>> > >>>>>>>>>>> > >>>>>>>>>>>> On Aug 30, 2016, at 8:48 PM, Cube Central < > >> cubecentral@gmail.com <javascript:;> > >>>> <javascript:;> > >>>>>> <javascript:;>> > >>>>>>>>>> wrote: > >>>>>>>>>>>> > >>>>>>>>>>>> I would be interested, I think. Planning ahead for if the > one I > >>>>>> have > >>>>>>>>>> for my Thunderbolt fails, I guess. Are there different models > or > >>>>>> would a > >>>>>>>>>> photo of the input ports on mine be useful? > >>>>>>>>>>> > >>>>>>>>>>> Actually, what I had in mind is to just put a SIP4 header on > the > >>>>>> board > >>>>>>>>>> for the output and people could wire the “last mile” themselves. > >> The > >>>>>> input > >>>>>>>>>> is a 2.1mm barrel connector. You use whatever 15W 12VDC wall > wart > >> is > >>>>>> handy > >>>>>>>>>> and plug it right in. > >>>>>>>>>>> > >>>>>>>>>>> What it really amounts to is that you get +12 volts directly > from > >>>> the > >>>>>>>>>> input, then there’s a buck converter to drop the +12 down to +5 > >> and > >>>> an > >>>>>>>>>> inverter to generate -12 from the +12. Those 3 voltages, plus a > >>>>>> ground go > >>>>>>>>>> to the SIP4. > >>>>>>>>>>> > >>>>>>>>>>> So it’s just two switching power supplies to turn a +12 volt > only > >>>>>>>>>> supply into the three-way that the Thunderbolt wants. > >>>>>>>>>>> > >>>>>>>>>>> It’d be good for around 1500 mA @ 5V and around 50 mA @ -12 > (the > >>>> +12 > >>>>>>>>>> spec is whatever is left from the source supply’s power spec) - > >> more > >>>>>> than > >>>>>>>>>> enough for a Thunderbolt. Probably enough for a hard disk or a > >>>>>> smallish PC. > >>>>>>>>>> _______________________________________________ > >>>>>>>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> > <javascript:;> > >>>> <javascript:;> > >>>>>>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/m > >>>>>>>>>> ailman/listinfo/time-nuts > >>>>>>>>>> and follow the instructions there. > >>>>>>>>>> > >>>>>>>>> > >>>>>>>>> > >>>>>>>> > >>>>>>> <7912_1PLC.png><7912_AllanDeviation.png><7912_ > >>>>>> TempCo.png>_______________________________________________ > >>>>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> > <javascript:;> > >>>> <javascript:;> > >>>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ > >>>>>> mailman/listinfo/time-nuts > >>>>>>> and follow the instructions there. > >>>>>> > >>>>>> _______________________________________________ > >>>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> > <javascript:;> > >>>> <javascript:;> > >>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ > >>>>>> mailman/listinfo/time-nuts > >>>>>> and follow the instructions there. > >>>>>> > >>>>> _______________________________________________ > >>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> > <javascript:;> > >>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ > >>>> mailman/listinfo/time-nuts > >>>>> and follow the instructions there. > >>>> > >>>> _______________________________________________ > >>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> > <javascript:;> > >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ > >>>> mailman/listinfo/time-nuts > >>>> and follow the instructions there. > >>>> > >>> _______________________________________________ > >>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> > >>> To unsubscribe, go to https://www.febo.com/cgi-bin/ > >> mailman/listinfo/time-nuts > >>> and follow the instructions there. > >> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@febo.com <javascript:;> > >> To unsubscribe, go to https://www.febo.com/cgi-bin/ > >> mailman/listinfo/time-nuts > >> and follow the instructions there. > >> > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com <javascript:;> > > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com <javascript:;> > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > and follow the instructions there. >
NS
Nick Sayer
Sat, Oct 22, 2016 8:45 PM

At this point, I have three prototype boards that are functional as far as I can tell. Unfortunately I don’t have a better reference which would allow me to make comparisons.

If someone with a better reference could make a proper comparison, I’d be happy to send them a prototype.

On Oct 22, 2016, at 1:15 PM, Scott Stobbe scott.j.stobbe@gmail.com wrote:

I think we are coming up to the noise floor here. We have identified a
potential signal A from a 7912 (likely similar to a 7812, no guarentee).
The remaining question is if you apply signal A to the power pin of a
thunderbolt what is signal B added to the output. We know where we want
signal B to be, below the noise. Knowing the PSSR for each of the power
inputs would be one way to answer this question.

On Saturday, 22 October 2016, Bob Camp <kb8tq@n1k.org mailto:kb8tq@n1k.org> wrote:

Hi

There is an enormous difference between regulation and it’s impact on
stability (which is what this drifted off into) and PSRR and it’s impact on
phase noise (which got lost at the tread moved on). PSRR does indeed
matter, but a fairly simple linear regulator (or pair of cheap ones or
coil plus
pair) can take care of that.

The EFC reference in the TBolt comes from the TBolt board and not from the
OCXO.
It is not in any way connected to the -12 V supply line.

Bob

On Oct 22, 2016, at 12:41 PM, Scott Stobbe <scott.j.stobbe@gmail.com

javascript:;> wrote:

Bob, this is good data and insights thanks for taking the time to share.
Ultimately the question Nick is hoping to answer is what is the point of
diminishing returns for voltage regulation. I think there are plenty of
folks on this list that have shared data suggesting a switching regulator
which meets these specifications:

Table 3-2 Power Specifications (Board-only)
+12v +/-10% max current 750 mA
+5v +/-5% max current 400 mA
-12v +/-10% max current 10 mA
Ripple
+5v 50 mV peak-to-peak, 15 mVrms
+/-12v 75 mV peak-to-peak, 20 mVrms

definitely has a degrading impact on the 10MHz phase noise.

I would happily agree that PSRR of an opamp buffering the EFC line is not
going to have a significant impact on performance with a jellybean LDO.

Do you know if the +- 5V reference is derived on board or is it from an

EFC

reference provided by the ocxo?

On Sat, Oct 22, 2016 at 11:31 AM, Bob Camp <kb8tq@n1k.org mailto:kb8tq@n1k.org javascript:;>

wrote:

Hi

The -12 V line is not the reference for the EFC. It is one supply into
the op amp circuit
that drives the EFC. Since the EFC is +/- 5V, there must be both a
positive and
a negative supply into the driver circuit for full output swing. The -12
can be anything between about
-13 and -7 without significantly impacting the function of the device.

In

fact, people
with EFC voltages that are slightly positive have accidentally proven

that

the
TBolt will work (lock up and function correctly) with zero volts on the
-12V line.

Bob

On Oct 22, 2016, at 10:24 AM, Scott Stobbe <scott.j.stobbe@gmail.com mailto:scott.j.stobbe@gmail.com

wrote:

It all depends what the -12V rail is for, some have said it directly
references the EFC dac. I would hope an ocxo would have a better tuning
gain on its efc pin than supply pin but maybe that's not always true.

On Saturday, 22 October 2016, Bob Camp <kb8tq@n1k.org mailto:kb8tq@n1k.org javascript:;>

wrote:

Hi

The +12 goes straight to the OCXO. All OCXO’s have a voltage

sensitivity.

That sensitivity is much higher than voltage sensitivity is much

higher

than
what you see on the other two supply pins.

Bob

On Oct 21, 2016, at 11:06 PM, Scott Stobbe <scott.j.stobbe@gmail.com mailto:scott.j.stobbe@gmail.com

javascript:;> wrote:

Interesting, I would have thought that the +12V input would be

extremely

well regulated since its shared with the oven heater, I*R drops are

going

to show up every where, if your looking for uV levels of stability.
Just a connector has milliohms of contact resistance, let alone

routing

and

cables...

On Friday, 21 October 2016, Bob Camp <kb8tq@n1k.org mailto:kb8tq@n1k.org javascript:;

wrote:

Hi

On Oct 21, 2016, at 7:44 PM, Scott Stobbe <

javascript:;> wrote:

A little more data on the 7912.

The first plot shows the tempCo of the 7912 measured with ambient
temperature swings "7912_TempCo.png". Which is -150 ppm/degC.

The second plot is off a 7912 logged for an hour or so,

"7912_1PLC.png",

nothing too interesting here. However the environmental temperature

swing

of about 1 degC/hour is pretty conservative for a DUT sitting in

free

air.

Finally, an allan devation plot looking at the normalized stability

of

a

7912 regulator "7912_AllanDeviation.png". Interestingly here, is,

how

quick

a 15 mK/min temperature swing shoots above the 1/f floor, it's a

matter

of

seconds.

Now if your PSRR is 1 ppb/V or better, then all of this is

comfortably

below the intrinsic noise of a thunderbolt.

The main point is that the internal tempco of the TBolt it’s self is

much

larger than
the issues surrounding the power supply pins. The +12 is the only

one

that

is
sensitive enough to voltage (change in frequency vs voltage) to

contribute

to any
significant way to the overall stability.

Bob

On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe <

wrote:

Nick had mention that the -12V rail on the thunderbolt has the

poorest

PSRR with respect to frequency output, so I first took a look at

the

venerable 7912.

The first data-set was taken with a -13.5 VDC input. Attached is

the

0.1

Hz to 10 Hz noise of an essentially quiescently loaded 7912, only

a

10k

resistor was added as load for preliminary evaluation. With a 60

dB

preamp

the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band

noise

is

15 uVrms, which is about 1.3 ppm rms of the DC mean.

In allan deviation terms, a quiescently loaded 7912 has a spot

noise

of

7

uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580

ppb/rtHz.

Equivalently speaking, the flicker noise floor of an allan

deviation

plot

would be sqrt(2*ln(2)) that figure to be 6.8E-7.

Assuming a thunderbolt should be achieving 1/f floor of around

1E-12,

it

would need a PSRR of at least 1 ppm/V. I'm sure someone has gone

to

the

trouble of actually measuring it.

So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't

terrible

with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms,

only

one

order of magnitude off.

The bad side of a 7912 is in long-term stability and tempCo, the

sample

I

tested had at least a 150 ppm/degC tempCo, which is going to put a

serious

lump/bump in the 10s tau to gps crossover point on an allan

deviation

plot.

On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe <

wrote:

I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm
interested to see the 1/f noise of a classic regulator, what load

current

do you expect? I can bias a 7805 for the same load and measure

the

0.1

to

10 Hz noise.

Also if you have a digital scope without a very good builtin FFT,

octave

would be one solution.

On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts <
time-nuts@febo.com mailto:time-nuts@febo.com javascript:; javascript:;

javascript:;> wrote:

Just an update. I’ve built the second prototype board (I skipped

over

the first design), and it’s powering my tbolt right now.

The design calls for 15v in (though it would also work with

13.8v).

The

+12 output comes from a D2PAK 7812. For +5, there is an AP1509

buck

converter to make around 6.5 volts, then a DPAK 7805. For -12,

there

is an

MC34063 configured as an inverter to make around -13.75 volts

and

then a

DPAK 7912.

Steady-state, the system appears to be working just fine. The

AP1509’s

inductor and the D2PAK 7812 are just warm to the touch.

I checked for noise and ripple on the outputs and it’s somewhere

around

±2 mV or so generally. From what I can see on the scope, there’s

no

ripple

  • it’s all high frequency noise. I am not absolutely certain

that

the

noise

measurement represents real noise or the limits of my measuring

ability.

I’m just using the scope probes the scope came with, and 2

mV/div

is

its

lowest range.

I haven’t compared the noise with the ex laptop supply that I

was

using

before, but I’d have to believe it’s cleaner. I don’t really

have

a

way to

check the oscillator’s before and after ADEV. My only other

reference

is an

FE5680A, and I think the thunderbolt’s going to be far better at

lower tau

(where this all matters).

I know also that ±2 mV is still one and perhaps two orders of

magnitude

higher than some have called for. But before I attempt to reduce

the

noise

further, I’d like to know that there are real gains to be had.

Would

someone with a Thunderbolt and better output noise measuring

wherewithal be

willing to take a prototype and compare it with something that

does

have µV

levels of noise and ripple so I can get an idea of what there is

to

gain?

If you like, you can make such comparisons public - no secrets

here.

On Aug 30, 2016, at 10:37 PM, Nick Sayer <nsayer@kfu.com mailto:nsayer@kfu.com

javascript:;> wrote:

On Aug 30, 2016, at 8:48 PM, Cube Central <

wrote:

I would be interested, I think.  Planning ahead for if the

one I

have

for my Thunderbolt fails, I guess.  Are there different models

or

would a

photo of the input ports on mine be useful?

Actually, what I had in mind is to just put a SIP4 header on

the

board

for the output and people could wire the “last mile” themselves.

The

input

is a 2.1mm barrel connector. You use whatever 15W 12VDC wall

wart

is

handy

and plug it right in.

What it really amounts to is that you get +12 volts directly

from

the

input, then there’s a buck converter to drop the +12 down to +5

and

an

inverter to generate -12 from the +12. Those 3 voltages, plus a

ground go

to the SIP4.

So it’s just two switching power supplies to turn a +12 volt

only

supply into the three-way that the Thunderbolt wants.

It’d be good for around 1500 mA @ 5V and around 50 mA @ -12

(the

+12

spec is whatever is left from the source supply’s power spec) -

more

than

enough for a Thunderbolt. Probably enough for a hard disk or a

smallish PC.


time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

<7912_1PLC.png><7912_AllanDeviation.png><7912_

TempCo.png>_______________________________________________

time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;

To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

At this point, I have three prototype boards that are functional as far as I can tell. Unfortunately I don’t have a better reference which would allow me to make comparisons. If someone with a better reference could make a proper comparison, I’d be happy to send them a prototype. > On Oct 22, 2016, at 1:15 PM, Scott Stobbe <scott.j.stobbe@gmail.com> wrote: > > I think we are coming up to the noise floor here. We have identified a > potential signal A from a 7912 (likely similar to a 7812, no guarentee). > The remaining question is if you apply signal A to the power pin of a > thunderbolt what is signal B added to the output. We know where we want > signal B to be, below the noise. Knowing the PSSR for each of the power > inputs would be one way to answer this question. > > On Saturday, 22 October 2016, Bob Camp <kb8tq@n1k.org <mailto:kb8tq@n1k.org>> wrote: > >> Hi >> >> There is an *enormous* difference between regulation and it’s impact on >> stability (which is what this drifted off into) and PSRR and it’s impact on >> phase noise (which got lost at the tread moved on). PSRR does indeed >> matter, but a fairly simple linear regulator (or pair of cheap ones or >> coil plus >> pair) can take care of that. >> >> The EFC reference in the TBolt comes from the TBolt board and not from the >> OCXO. >> It is not in any way connected to the -12 V supply line. >> >> Bob >> >> >> >>> On Oct 22, 2016, at 12:41 PM, Scott Stobbe <scott.j.stobbe@gmail.com >> <javascript:;>> wrote: >>> >>> Bob, this is good data and insights thanks for taking the time to share. >>> Ultimately the question Nick is hoping to answer is what is the point of >>> diminishing returns for voltage regulation. I think there are plenty of >>> folks on this list that have shared data suggesting a switching regulator >>> which meets these specifications: >>> >>> Table 3-2 Power Specifications (Board-only) >>> +12v +/-10% max current 750 mA >>> +5v +/-5% max current 400 mA >>> -12v +/-10% max current 10 mA >>> Ripple >>> +5v 50 mV peak-to-peak, 15 mVrms >>> +/-12v 75 mV peak-to-peak, 20 mVrms >>> >>> definitely has a degrading impact on the 10MHz phase noise. >>> >>> I would happily agree that PSRR of an opamp buffering the EFC line is not >>> going to have a significant impact on performance with a jellybean LDO. >>> >>> Do you know if the +- 5V reference is derived on board or is it from an >> EFC >>> reference provided by the ocxo? >>> >>> On Sat, Oct 22, 2016 at 11:31 AM, Bob Camp <kb8tq@n1k.org <mailto:kb8tq@n1k.org> <javascript:;>> >> wrote: >>> >>>> Hi >>>> >>>> The -12 V line is not the reference for the EFC. It *is* one supply into >>>> the op amp circuit >>>> that drives the EFC. Since the EFC is +/- 5V, there must be both a >>>> positive and >>>> a negative supply into the driver circuit for full output swing. The -12 >>>> can be anything between about >>>> -13 and -7 without significantly impacting the function of the device. >> In >>>> fact, people >>>> with EFC voltages that are slightly positive have accidentally proven >> that >>>> the >>>> TBolt will work (lock up and function correctly) with zero volts on the >>>> -12V line. >>>> >>>> Bob >>>> >>>>> On Oct 22, 2016, at 10:24 AM, Scott Stobbe <scott.j.stobbe@gmail.com <mailto:scott.j.stobbe@gmail.com> >> <javascript:;>> >>>> wrote: >>>>> >>>>> It all depends what the -12V rail is for, some have said it directly >>>>> references the EFC dac. I would hope an ocxo would have a better tuning >>>>> gain on its efc pin than supply pin but maybe that's not always true. >>>>> >>>>> On Saturday, 22 October 2016, Bob Camp <kb8tq@n1k.org <mailto:kb8tq@n1k.org> <javascript:;>> >> wrote: >>>>> >>>>>> Hi >>>>>> >>>>>> The +12 goes straight to the OCXO. All OCXO’s have a voltage >>>> sensitivity. >>>>>> That sensitivity is much higher than voltage sensitivity is much >> higher >>>>>> than >>>>>> what you see on the other two supply pins. >>>>>> >>>>>> Bob >>>>>> >>>>>>> On Oct 21, 2016, at 11:06 PM, Scott Stobbe <scott.j.stobbe@gmail.com <mailto:scott.j.stobbe@gmail.com> >> <javascript:;> >>>>>> <javascript:;>> wrote: >>>>>>> >>>>>>> Interesting, I would have thought that the +12V input would be >>>> extremely >>>>>>> well regulated since its shared with the oven heater, I*R drops are >>>> going >>>>>>> to show up every where, if your looking for uV levels of stability. >>>>>>> Just a connector has milliohms of contact resistance, let alone >> routing >>>>>> and >>>>>>> cables... >>>>>>> >>>>>>> On Friday, 21 October 2016, Bob Camp <kb8tq@n1k.org <mailto:kb8tq@n1k.org> <javascript:;> >> <javascript:;>> >>>>>> wrote: >>>>>>> >>>>>>>> Hi >>>>>>>> >>>>>>>>> On Oct 21, 2016, at 7:44 PM, Scott Stobbe < >> scott.j.stobbe@gmail.com <mailto:scott.j.stobbe@gmail.com> <javascript:;> >>>>>> <javascript:;> >>>>>>>> <javascript:;>> wrote: >>>>>>>>> >>>>>>>>> A little more data on the 7912. >>>>>>>>> >>>>>>>>> The first plot shows the tempCo of the 7912 measured with ambient >>>>>>>>> temperature swings "7912_TempCo.png". Which is -150 ppm/degC. >>>>>>>>> >>>>>>>>> The second plot is off a 7912 logged for an hour or so, >>>>>> "7912_1PLC.png", >>>>>>>>> nothing too interesting here. However the environmental temperature >>>>>> swing >>>>>>>>> of about 1 degC/hour is pretty conservative for a DUT sitting in >> free >>>>>>>> air. >>>>>>>>> >>>>>>>>> Finally, an allan devation plot looking at the normalized stability >>>> of >>>>>> a >>>>>>>>> 7912 regulator "7912_AllanDeviation.png". Interestingly here, is, >> how >>>>>>>> quick >>>>>>>>> a 15 mK/min temperature swing shoots above the 1/f floor, it's a >>>> matter >>>>>>>> of >>>>>>>>> seconds. >>>>>>>>> >>>>>>>>> Now if your PSRR is 1 ppb/V or better, then all of this is >>>> comfortably >>>>>>>>> below the intrinsic noise of a thunderbolt. >>>>>>>> >>>>>>>> The main point is that the internal tempco of the TBolt it’s self is >>>>>> much >>>>>>>> larger than >>>>>>>> the issues surrounding the power supply pins. The +12 is the only >> one >>>>>> that >>>>>>>> is >>>>>>>> sensitive enough to voltage (change in frequency vs voltage) to >>>>>> contribute >>>>>>>> to any >>>>>>>> significant way to the overall stability. >>>>>>>> >>>>>>>> Bob >>>>>>>> >>>>>>>>> >>>>>>>>> On Fri, Oct 21, 2016 at 12:20 AM, Scott Stobbe < >>>>>> scott.j.stobbe@gmail.com <mailto:scott.j.stobbe@gmail.com> <javascript:;> <javascript:;> >>>>>>>> <javascript:;>> >>>>>>>>> wrote: >>>>>>>>> >>>>>>>>>> Nick had mention that the -12V rail on the thunderbolt has the >>>> poorest >>>>>>>>>> PSRR with respect to frequency output, so I first took a look at >> the >>>>>>>>>> venerable 7912. >>>>>>>>>> >>>>>>>>>> The first data-set was taken with a -13.5 VDC input. Attached is >> the >>>>>> 0.1 >>>>>>>>>> Hz to 10 Hz noise of an essentially quiescently loaded 7912, only >> a >>>>>> 10k >>>>>>>>>> resistor was added as load for preliminary evaluation. With a 60 >> dB >>>>>>>> preamp >>>>>>>>>> the scale of the scope plot is 20 uV/div. The 0.1Hz to 10Hz band >>>> noise >>>>>>>> is >>>>>>>>>> 15 uVrms, which is about 1.3 ppm rms of the DC mean. >>>>>>>>>> >>>>>>>>>> In allan deviation terms, a quiescently loaded 7912 has a spot >> noise >>>>>> of >>>>>>>> 7 >>>>>>>>>> uV/rtHz at 1 Hz (on the 1/f slope), normalized that's 580 >> ppb/rtHz. >>>>>>>>>> Equivalently speaking, the flicker noise floor of an allan >> deviation >>>>>>>> plot >>>>>>>>>> would be sqrt(2*ln(2)) that figure to be 6.8E-7. >>>>>>>>>> >>>>>>>>>> Assuming a thunderbolt should be achieving 1/f floor of around >>>> 1E-12, >>>>>> it >>>>>>>>>> would need a PSRR of at least 1 ppm/V. I'm sure someone has gone >> to >>>>>> the >>>>>>>>>> trouble of actually measuring it. >>>>>>>>>> >>>>>>>>>> So from a 0.1 Hz to 10 Hz noise standpoint, the 7912 isn't >> terrible >>>>>>>>>> with 1.3 ppm rms noise, considering an LM399 is about 0.1 ppm rms, >>>>>> only >>>>>>>> one >>>>>>>>>> order of magnitude off. >>>>>>>>>> >>>>>>>>>> The bad side of a 7912 is in long-term stability and tempCo, the >>>>>> sample >>>>>>>> I >>>>>>>>>> tested had at least a 150 ppm/degC tempCo, which is going to put a >>>>>>>> serious >>>>>>>>>> lump/bump in the 10s tau to gps crossover point on an allan >>>> deviation >>>>>>>> plot. >>>>>>>>>> >>>>>>>>>> >>>>>>>>>> >>>>>>>>>> On Tue, Oct 18, 2016 at 3:05 PM, Scott Stobbe < >>>>>> scott.j.stobbe@gmail.com <mailto:scott.j.stobbe@gmail.com> <javascript:;> <javascript:;> >>>>>>>> <javascript:;>> >>>>>>>>>> wrote: >>>>>>>>>> >>>>>>>>>>> I'm sure I have some 7805s lying around, maybe a 7812/7912. I'm >>>>>>>>>>> interested to see the 1/f noise of a classic regulator, what load >>>>>>>> current >>>>>>>>>>> do you expect? I can bias a 7805 for the same load and measure >> the >>>>>> 0.1 >>>>>>>> to >>>>>>>>>>> 10 Hz noise. >>>>>>>>>>> >>>>>>>>>>> Also if you have a digital scope without a very good builtin FFT, >>>>>>>> octave >>>>>>>>>>> would be one solution. >>>>>>>>>>> >>>>>>>>>>> On Tue, Oct 18, 2016 at 10:46 AM, Nick Sayer via time-nuts < >>>>>>>>>>> time-nuts@febo.com <mailto:time-nuts@febo.com> <javascript:;> <javascript:;> >> <javascript:;>> wrote: >>>>>>>>>>> >>>>>>>>>>>> Just an update. I’ve built the second prototype board (I skipped >>>>>> over >>>>>>>>>>>> the first design), and it’s powering my tbolt right now. >>>>>>>>>>>> >>>>>>>>>>>> The design calls for 15v in (though it would also work with >>>> 13.8v). >>>>>>>> The >>>>>>>>>>>> +12 output comes from a D2PAK 7812. For +5, there is an AP1509 >>>> buck >>>>>>>>>>>> converter to make around 6.5 volts, then a DPAK 7805. For -12, >>>> there >>>>>>>> is an >>>>>>>>>>>> MC34063 configured as an inverter to make around -13.75 volts >> and >>>>>>>> then a >>>>>>>>>>>> DPAK 7912. >>>>>>>>>>>> >>>>>>>>>>>> Steady-state, the system appears to be working just fine. The >>>>>> AP1509’s >>>>>>>>>>>> inductor and the D2PAK 7812 are just warm to the touch. >>>>>>>>>>>> >>>>>>>>>>>> I checked for noise and ripple on the outputs and it’s somewhere >>>>>>>> around >>>>>>>>>>>> ±2 mV or so generally. From what I can see on the scope, there’s >>>> no >>>>>>>> ripple >>>>>>>>>>>> - it’s all high frequency noise. I am not absolutely certain >> that >>>>>> the >>>>>>>> noise >>>>>>>>>>>> measurement represents real noise or the limits of my measuring >>>>>>>> ability. >>>>>>>>>>>> I’m just using the scope probes the scope came with, and 2 >> mV/div >>>> is >>>>>>>> its >>>>>>>>>>>> lowest range. >>>>>>>>>>>> >>>>>>>>>>>> I haven’t compared the noise with the ex laptop supply that I >> was >>>>>>>> using >>>>>>>>>>>> before, but I’d have to believe it’s cleaner. I don’t really >> have >>>> a >>>>>>>> way to >>>>>>>>>>>> check the oscillator’s before and after ADEV. My only other >>>>>> reference >>>>>>>> is an >>>>>>>>>>>> FE5680A, and I think the thunderbolt’s going to be far better at >>>>>>>> lower tau >>>>>>>>>>>> (where this all matters). >>>>>>>>>>>> >>>>>>>>>>>> I know also that ±2 mV is still one and perhaps two orders of >>>>>>>> magnitude >>>>>>>>>>>> higher than some have called for. But before I attempt to reduce >>>> the >>>>>>>> noise >>>>>>>>>>>> further, I’d like to know that there are real gains to be had. >>>> Would >>>>>>>>>>>> someone with a Thunderbolt and better output noise measuring >>>>>>>> wherewithal be >>>>>>>>>>>> willing to take a prototype and compare it with something that >>>> does >>>>>>>> have µV >>>>>>>>>>>> levels of noise and ripple so I can get an idea of what there is >>>> to >>>>>>>> gain? >>>>>>>>>>>> If you like, you can make such comparisons public - no secrets >>>> here. >>>>>>>>>>>> >>>>>>>>>>>>> On Aug 30, 2016, at 10:37 PM, Nick Sayer <nsayer@kfu.com <mailto:nsayer@kfu.com> >> <javascript:;> >>>>>> <javascript:;> >>>>>>>> <javascript:;>> wrote: >>>>>>>>>>>>> >>>>>>>>>>>>> >>>>>>>>>>>>>> On Aug 30, 2016, at 8:48 PM, Cube Central < >>>> cubecentral@gmail.com <mailto:cubecentral@gmail.com> <javascript:;> >>>>>> <javascript:;> >>>>>>>> <javascript:;>> >>>>>>>>>>>> wrote: >>>>>>>>>>>>>> >>>>>>>>>>>>>> I would be interested, I think. Planning ahead for if the >> one I >>>>>>>> have >>>>>>>>>>>> for my Thunderbolt fails, I guess. Are there different models >> or >>>>>>>> would a >>>>>>>>>>>> photo of the input ports on mine be useful? >>>>>>>>>>>>> >>>>>>>>>>>>> Actually, what I had in mind is to just put a SIP4 header on >> the >>>>>>>> board >>>>>>>>>>>> for the output and people could wire the “last mile” themselves. >>>> The >>>>>>>> input >>>>>>>>>>>> is a 2.1mm barrel connector. You use whatever 15W 12VDC wall >> wart >>>> is >>>>>>>> handy >>>>>>>>>>>> and plug it right in. >>>>>>>>>>>>> >>>>>>>>>>>>> What it really amounts to is that you get +12 volts directly >> from >>>>>> the >>>>>>>>>>>> input, then there’s a buck converter to drop the +12 down to +5 >>>> and >>>>>> an >>>>>>>>>>>> inverter to generate -12 from the +12. Those 3 voltages, plus a >>>>>>>> ground go >>>>>>>>>>>> to the SIP4. >>>>>>>>>>>>> >>>>>>>>>>>>> So it’s just two switching power supplies to turn a +12 volt >> only >>>>>>>>>>>> supply into the three-way that the Thunderbolt wants. >>>>>>>>>>>>> >>>>>>>>>>>>> It’d be good for around 1500 mA @ 5V and around 50 mA @ -12 >> (the >>>>>> +12 >>>>>>>>>>>> spec is whatever is left from the source supply’s power spec) - >>>> more >>>>>>>> than >>>>>>>>>>>> enough for a Thunderbolt. Probably enough for a hard disk or a >>>>>>>> smallish PC. >>>>>>>>>>>> _______________________________________________ >>>>>>>>>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >> <javascript:;> >>>>>> <javascript:;> >>>>>>>>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/m >>>>>>>>>>>> ailman/listinfo/time-nuts >>>>>>>>>>>> and follow the instructions there. >>>>>>>>>>>> >>>>>>>>>>> >>>>>>>>>>> >>>>>>>>>> >>>>>>>>> <7912_1PLC.png><7912_AllanDeviation.png><7912_ >>>>>>>> TempCo.png>_______________________________________________ >>>>>>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >> <javascript:;> >>>>>> <javascript:;> >>>>>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>>>>>>> mailman/listinfo/time-nuts >>>>>>>>> and follow the instructions there. >>>>>>>> >>>>>>>> _______________________________________________ >>>>>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >> <javascript:;> >>>>>> <javascript:;> >>>>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>>>>>>> mailman/listinfo/time-nuts >>>>>>>> and follow the instructions there. >>>>>>>> >>>>>>> _______________________________________________ >>>>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >> <javascript:;> >>>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>>>>> mailman/listinfo/time-nuts >>>>>>> and follow the instructions there. >>>>>> >>>>>> _______________________________________________ >>>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >> <javascript:;> >>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>>>>> mailman/listinfo/time-nuts >>>>>> and follow the instructions there. >>>>>> >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>>> mailman/listinfo/time-nuts >>>>> and follow the instructions there. >>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>>> mailman/listinfo/time-nuts >>>> and follow the instructions there. >>>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >> mailman/listinfo/time-nuts >>> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >> To unsubscribe, go to https://www.febo.com/cgi-bin/ >> mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com <mailto:time-nuts@febo.com> > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts <https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts> > and follow the instructions there.