On 2016-09-26 10:00 AM, Christopher Hoover ch@murgatroid.com wrote:
You might be able to slide something like a feeler guage down between the
oven and the rubber blanket to break the oscillator free. The oven on mine
is a plain metal cylinder. This way, the rubber sheet should protect the
Dewar from your feeler guage. On mine, the mounting bolts for the 2N3792
transistor both have ground lugs. I think I see them on yours. You could
hook something through the ground lugs and use that to pull the oscillator
out of the rubber sheet and then remove the sheet later.
Thanks Ed,
I think the rubber sheet on mine is against metal. I haven't yet seen the
glass dewar.
The adhesion is huge.
Do you know if the holes opposite the 2N3792 are threaded? If they are, I
might try running the screws out and using those holes with longer screws
as my pull points. I can't pull on the lugs hard enough -- I've tried.
-christopher.
73 de AI6KG
Yes, you have seen the Dewar. The silvery ring that's outside the
rubber is the top of the Dewar. What you have to do is unstick and
unfold the rubber starting from the open area in the center. Work your
way outward. The rubber is only 2 or 3 mm thick. Once you completely
clear the rubber out of the way, you'll see the edge of the oven. The
TO-3 transistor is mounted on top of the oven assembly. Once you can
see the edge, you have to slide something like a long feeler gauge down
along the edge of the oven to break it free from the rubber. Work your
way all around the oven. It's about 85 mm long. It'll still be stuck
on the bottom, but you might be able to pull it free.
When I took mine apart, I ended up tearing off all the rubber at the top
and then cutting out that ring of hard foam to get at the Dewar so I
could smash it more. I'm guessing you'd rather not do that! :) But
sacrificing the rubber on the top might be okay, if you have to.
Sorry, but I don't know if the mounting holes for the transistor are
threaded or not. In any case, since the oven and Dewar are bonded to
the rubber, you're pulling on the Dewar when you pull on the oven. Not
a good plan until you break the oven free from the rubber. Those Dewars
are built in a rather fragile manner. Your typical home Thermos is much
more robust.
Ed
Back when I was going to work on mine, I was thinking of
prying the rubber away from the aluminum oven with something
like a feeler gauge, but also using some naptha (lighter fluid)
to help release any adhesive... I didn't get around to doing
it, but that was the way I was going to progress.
-Chuck Harris
Ed Palmer wrote:
On 2016-09-26 10:00 AM, Christopher Hoover ch@murgatroid.com wrote:
You might be able to slide something like a feeler guage down between the
oven and the rubber blanket to break the oscillator free. The oven on mine
is a plain metal cylinder. This way, the rubber sheet should protect the
Dewar from your feeler guage. On mine, the mounting bolts for the 2N3792
transistor both have ground lugs. I think I see them on yours. You could
hook something through the ground lugs and use that to pull the oscillator
out of the rubber sheet and then remove the sheet later.
Thanks Ed,
I think the rubber sheet on mine is against metal. I haven't yet seen the
glass dewar.
The adhesion is huge.
Do you know if the holes opposite the 2N3792 are threaded? If they are, I
might try running the screws out and using those holes with longer screws
as my pull points. I can't pull on the lugs hard enough -- I've tried.
-christopher.
73 de AI6KG
Yes, you have seen the Dewar. The silvery ring that's outside the rubber is the top
of the Dewar. What you have to do is unstick and unfold the rubber starting from the
open area in the center. Work your way outward. The rubber is only 2 or 3 mm
thick. Once you completely clear the rubber out of the way, you'll see the edge of
the oven. The TO-3 transistor is mounted on top of the oven assembly. Once you can
see the edge, you have to slide something like a long feeler gauge down along the
edge of the oven to break it free from the rubber. Work your way all around the
oven. It's about 85 mm long. It'll still be stuck on the bottom, but you might be
able to pull it free.
When I took mine apart, I ended up tearing off all the rubber at the top and then
cutting out that ring of hard foam to get at the Dewar so I could smash it more. I'm
guessing you'd rather not do that! :) But sacrificing the rubber on the top might be
okay, if you have to.
Sorry, but I don't know if the mounting holes for the transistor are threaded or
not. In any case, since the oven and Dewar are bonded to the rubber, you're pulling
on the Dewar when you pull on the oven. Not a good plan until you break the oven
free from the rubber. Those Dewars are built in a rather fragile manner. Your
typical home Thermos is much more robust.
Ed
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
tl;dr: I've made some progress and have the1200 oscillator core out of the
dewar:
https://goo.gl/photos/SDHtvgFmftQq6vYJA
I got some 5 thou brass stock and worked it between the rubber sheet and
the dewar.
That seemed to help but it was insufficient free things up -- I busted off
the unused solder lug trying to pull the core out with it.
Having not a lot to lose, I took a chance that the screws going into the
TO-23 went into threaded holes (rather than being clearance holes with nuts
inside). This was indeed the case.
With two 6-32 threaded rods into the TO-23 threaded holes and and an
appropriately machined piece of mild steel bar stock suspended across the
case https://goo.gl/photos/1pfiN2GX3WxYCSbg8, I was able to easily get
the oscillator core out of the dewar by evenly tightening the the nuts on
the bar. Really easily -- I might have been able to pull it out by just
pulling on the bar stock. I don't know if the shim stock shenanigans were
even needed.
Despite running out the three sloted screws on the "top" around the
circumference, I'm not into the inside yet. I don't have the right
thin-walled socket to remove the nuts at the opposite end.
I found an epoxy covered hole on the top. It is/was under the green blob
midway between 1 and 2 o'clock in this picture here
https://goo.gl/photos/iHbSbqwBiKD7NRfJ6.. There was something blue and
at this point crumbly underneath it. Not sure yet what, if anything, is
beyond all of that. I'm hoping for a trimmer cap. :-)
-- Christopher.
73 de AI6KG
On Mon, Sep 26, 2016 at 9:24 PM, Chuck Harris cfharris@erols.com wrote:
Back when I was going to work on mine, I was thinking of
prying the rubber away from the aluminum oven with something
like a feeler gauge, but also using some naptha (lighter fluid)
to help release any adhesive... I didn't get around to doing
it, but that was the way I was going to progress.
-Chuck Harris
Ed Palmer wrote:
On 2016-09-26 10:00 AM, Christopher Hoover ch@murgatroid.com wrote:
You might be able to slide something like a feeler guage down between
the
oven and the rubber blanket to break the oscillator free. The oven
on mine
is a plain metal cylinder. This way, the rubber sheet should protect
the
Dewar from your feeler guage. On mine, the mounting bolts for the
2N3792
transistor both have ground lugs. I think I see them on yours. You
could
hook something through the ground lugs and use that to pull the
oscillator
out of the rubber sheet and then remove the sheet later.
Thanks Ed,
I think the rubber sheet on mine is against metal. I haven't yet seen
the
glass dewar.
The adhesion is huge.
Do you know if the holes opposite the 2N3792 are threaded? If they
are, I
might try running the screws out and using those holes with longer
screws
as my pull points. I can't pull on the lugs hard enough -- I've
tried.
-christopher.
73 de AI6KG
Yes, you have seen the Dewar. The silvery ring that's outside the
rubber is the top
of the Dewar. What you have to do is unstick and unfold the rubber
starting from the
open area in the center. Work your way outward. The rubber is only 2
or 3 mm
thick. Once you completely clear the rubber out of the way, you'll see
the edge of
the oven. The TO-3 transistor is mounted on top of the oven assembly.
Once you can
see the edge, you have to slide something like a long feeler gauge down
along the
edge of the oven to break it free from the rubber. Work your way all
around the
oven. It's about 85 mm long. It'll still be stuck on the bottom, but
you might be
able to pull it free.
When I took mine apart, I ended up tearing off all the rubber at the top
and then
cutting out that ring of hard foam to get at the Dewar so I could smash
it more. I'm
guessing you'd rather not do that! :) But sacrificing the rubber on the
top might be
okay, if you have to.
Sorry, but I don't know if the mounting holes for the transistor are
threaded or
not. In any case, since the oven and Dewar are bonded to the rubber,
you're pulling
on the Dewar when you pull on the oven. Not a good plan until you break
the oven
free from the rubber. Those Dewars are built in a rather fragile
manner. Your
typical home Thermos is much more robust.
Ed
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.
Christopher wrote:
Despite running out the three sloted screws on the "top" around the
circumference, I'm not into the inside yet. I don't have the right
thin-walled socket to remove the nuts at the opposite end.
It looks like a standard 5/32" nut driver will work fine.
http://www.homedepot.com/p/203310440
Best regards,
Charles
Christopher,
Enjoyed the pictures.
You might want to look at these items on theBay.
381408412092
311736541103
I've had the same issue and broke down and bought a set of these small nut drivers.
Good luck.
Joe
-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Christopher Hoover
Sent: Saturday, November 19, 2016 11:45 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Inside of FT1200-100
tl;dr: I've made some progress and have the1200 oscillator core out of the
dewar:
https://goo.gl/photos/SDHtvgFmftQq6vYJA
I got some 5 thou brass stock and worked it between the rubber sheet and the dewar.
That seemed to help but it was insufficient free things up -- I busted off the unused solder lug trying to pull the core out with it.
Having not a lot to lose, I took a chance that the screws going into the
TO-23 went into threaded holes (rather than being clearance holes with nuts inside). This was indeed the case.
With two 6-32 threaded rods into the TO-23 threaded holes and and an appropriately machined piece of mild steel bar stock suspended across the case https://goo.gl/photos/1pfiN2GX3WxYCSbg8, I was able to easily get the oscillator core out of the dewar by evenly tightening the the nuts on
the bar. Really easily -- I might have been able to pull it out by just
pulling on the bar stock. I don't know if the shim stock shenanigans were even needed.
Despite running out the three sloted screws on the "top" around the
circumference, I'm not into the inside yet. I don't have the right
thin-walled socket to remove the nuts at the opposite end.
I found an epoxy covered hole on the top. It is/was under the green blob
midway between 1 and 2 o'clock in this picture here https://goo.gl/photos/iHbSbqwBiKD7NRfJ6.. There was something blue and
at this point crumbly underneath it. Not sure yet what, if anything, is
beyond all of that. I'm hoping for a trimmer cap. :-)
-- Christopher.
73 de AI6KG
On Mon, Sep 26, 2016 at 9:24 PM, Chuck Harris cfharris@erols.com wrote:
Back when I was going to work on mine, I was thinking of prying the
rubber away from the aluminum oven with something like a feeler gauge,
but also using some naptha (lighter fluid) to help release any
adhesive... I didn't get around to doing it, but that was the way I
was going to progress.
-Chuck Harris
Ed Palmer wrote:
On 2016-09-26 10:00 AM, Christopher Hoover ch@murgatroid.com wrote:
You might be able to slide something like a feeler guage down
between
the
oven and the rubber blanket to break the oscillator free. The
oven
on mine
is a plain metal cylinder. This way, the rubber sheet should
protect
the
Dewar from your feeler guage. On mine, the mounting bolts for
the
2N3792
transistor both have ground lugs. I think I see them on yours.
You
could
hook something through the ground lugs and use that to pull the
oscillator
out of the rubber sheet and then remove the sheet later.
Thanks Ed,
I think the rubber sheet on mine is against metal. I haven't yet
seen
the
glass dewar.
The adhesion is huge.
Do you know if the holes opposite the 2N3792 are threaded? If they
are, I
might try running the screws out and using those holes with longer
screws
as my pull points. I can't pull on the lugs hard enough -- I've
tried.
-christopher.
73 de AI6KG
Yes, you have seen the Dewar. The silvery ring that's outside the
rubber is the top
of the Dewar. What you have to do is unstick and unfold the rubber
starting from the
open area in the center. Work your way outward. The rubber is only
2
or 3 mm
thick. Once you completely clear the rubber out of the way, you'll
see
the edge of
the oven. The TO-3 transistor is mounted on top of the oven assembly.
Once you can
see the edge, you have to slide something like a long feeler gauge
down
along the
edge of the oven to break it free from the rubber. Work your way
all
around the
oven. It's about 85 mm long. It'll still be stuck on the bottom,
but
you might be
able to pull it free.
When I took mine apart, I ended up tearing off all the rubber at the
top
and then
cutting out that ring of hard foam to get at the Dewar so I could
smash
it more. I'm
guessing you'd rather not do that! :) But sacrificing the rubber on
the
top might be
okay, if you have to.
Sorry, but I don't know if the mounting holes for the transistor are
threaded or
not. In any case, since the oven and Dewar are bonded to the
rubber,
you're pulling
on the Dewar when you pull on the oven. Not a good plan until you
break
the oven
free from the rubber. Those Dewars are built in a rather fragile
manner. Your
typical home Thermos is much more robust.
Ed
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/ mailman/listinfo/time-nuts and follow
the instructions there.
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.