Amongst the stuff I recently acquired is a Fluke 332B (sn 4151).
I found the manual for the 332D on the Fluke website, but can't find one
specifically for the 332B apart from an incomplete US Army version.
I have some questions:
Does anyone know where I can find the matching manual
If not, how useful is the 332D manual?
It was last calibrated in 2010, and appears to be consistently low in
output (about 5mV at 100V, 25mV at 500V, 50mV at 1000V) after being on for
an hour. Is that the sort of error I should expect after that period?
I'm leery of the chopper amp, did anyone (Dr. Frank?) do a retrofit using
a modern chopper op-amp or similar?
Thanks
Dave
David:
The 332D and the 332B are identical except for the accuracy specs for
the stability of the output voltages.
332B spec for 10 Volts is +/- .002% (20 ppm) for 90 days while the 332D
spec for 10 Volts is +/- .001% (10 ppm) for 60 days. The monthly drift
specs are also slightly better on the D than the B.
I suspect that FLUKE simply aged and selected a more stable reference
device along with matching the resistors in the sampling string for less
drift. For all of that you paid US$2,595 for the 332B and US$2,995 for the
332D in 1973.
I seem to remember someone replacing the chopper assembly with something
a little more modern. Don't know where I saw that. I have a B and a D and
both seemed to be very stable with the existing chopper assembly.
One thing I did find is that the electrolitic caps have a high tendency
to go bad. Not the big ones on the raw supply but those on the various plug
in assemblies. I seem to remember that they were in a range of 1 ufd to 47
ufd or so. So check them all or better yet just recap. You will save
yourself a LOT of troubleshooting time in the long run. There is also a
discussion of this subject either on VoltNuts or EEVBLOG.
Bill
----- Original Message -----
From: "David C. Partridge" david.partridge@perdrix.co.uk
To: "'Discussion of precise voltage measurement'" volt-nuts@febo.com
Sent: Saturday, July 08, 2017 5:09 AM
Subject: [volt-nuts] Fluke 332B
Amongst the stuff I recently acquired is a Fluke 332B (sn 4151).
I found the manual for the 332D on the Fluke website, but can't find one
specifically for the 332B apart from an incomplete US Army version.
I have some questions:
Does anyone know where I can find the matching manual
If not, how useful is the 332D manual?
It was last calibrated in 2010, and appears to be consistently low in
output (about 5mV at 100V, 25mV at 500V, 50mV at 1000V) after being on for
an hour. Is that the sort of error I should expect after that period?
I'm leery of the chopper amp, did anyone (Dr. Frank?) do a retrofit
using
a modern chopper op-amp or similar?
Thanks
Dave
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
and follow the instructions there.
Hello Dave,
Am 08.07.2017 14:09, schrieb David C. Partridge:
You may want to have a look at this thread:
https://www.eevblog.com/forum/testgear/fluke-332baf-in-the-slaughterhouse/
http://www.friedrich-messtechnik.de/index.php/bauanleitungen/87-fluke335-332
(sorry, it is German)
I have created some PCBs based on the schematics shown on this website.
If you are interested, I can send you one for the shipping costs.
It is working well in my Fluke 3330B which is using the same design as
the 332.
Regards,
Andreas
On 08/07/2017 21:36, Andreas Bergmann wrote:
Hello Dave,
Am 08.07.2017 14:09, schrieb David C. Partridge:
You may want to have a look at this thread:
https://www.eevblog.com/forum/testgear/fluke-332baf-in-the-slaughterhouse/
http://www.friedrich-messtechnik.de/index.php/bauanleitungen/87-fluke335-332 (sorry, it is German)
I have created some PCBs based on the schematics shown on this website.
If you are interested, I can send you one for the shipping costs.
It is working well in my Fluke 3330B which is using the same design as the 332.
Regards,
Andreas
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
.
Hi Andreas,
I also have a 3330B which could do with an updated chopper PCB, how many spare boards do you have ( I only need 1)
Regards,
M K