Connecting the 10 MHz output to the Channel A input (on the 5334B) or the
frequency input (on the 5370B) and setting 50 Ohm termination and AC signals
but without any other input or external reference results in a frequency
that is off by about 0.1-0.2 Hz. Is this amount of offset expected or a sign
something is not right with the GPSDO?
Your GPSDO is more accurate than your counter. You are measuring the
frequency offset of the crystal in the counter rather than the GPSDO.
You may have to stand on your head to understand what I'm trying to say. It
will be obvious once you see it.
If/when you power cycle the GPSDO (without power cycling the counter) you can
watch it shift when it locks up to GPS. If you disconnect the antenna it
will stay unlocked and you can watch the crystal in the GPSDO drift as it
warms up.
If you unplug the counter for a while but leave the GPSDO powered on, you can
watch the crystal in the counter warm up. (Turning it off isn't good enough,
it keeps the crystal warm even when the switch says off.)
If you have a GPIB connection to the 5334B, you can get a few more digits by
asking it to count for 99 seconds. If you get that working and your lab is
not air conditioned, you can see the temperature effects on the 5334B.
--
These are my opinions. I hate spam.
On Fri, Aug 19, 2016 at 12:24 PM, Hal Murray hmurray@megapathdsl.net wrote:
listertim@gmail.com said:
Connecting the 10 MHz output to the Channel A input (on the 5334B) or the
frequency input (on the 5370B) and setting 50 Ohm termination and AC signals
but without any other input or external reference results in a frequency
that is off by about 0.1-0.2 Hz. Is this amount of offset expected or a sign
something is not right with the GPSDO?
Your GPSDO is more accurate than your counter. You are measuring the
frequency offset of the crystal in the counter rather than the GPSDO.
You may have to stand on your head to understand what I'm trying to say. It
will be obvious once you see it.
If/when you power cycle the GPSDO (without power cycling the counter) you can
watch it shift when it locks up to GPS. If you disconnect the antenna it
will stay unlocked and you can watch the crystal in the GPSDO drift as it
warms up.
If you unplug the counter for a while but leave the GPSDO powered on, you can
watch the crystal in the counter warm up. (Turning it off isn't good enough,
it keeps the crystal warm even when the switch says off.)
If you have a GPIB connection to the 5334B, you can get a few more digits by
asking it to count for 99 seconds. If you get that working and your lab is
not air conditioned, you can see the temperature effects on the 5334B.
Thanks to everyone for the replies, this definitely helps clear things
up. I had my suspicions that I wasn't measuring what I thought I was
as the resulting ADEV curve I was getting in Timelab (I forgot to
mention I have a Prologix GPIB->USB interface and have been using it
for logging/capture) had the '_/' (down, flat, up) shape that I
gather you get from OCXO's. I did also have a suspicion that this
could lead down a slippery slope of needing another GPSDO to compare
it with and then and then... ;-) (While I see there is a 5071A
available on ebay right now, those are somewhat out of my budget by
several orders of magnitude...)
Thanks,