Absolute phase

JH
Jerry Hancock
Fri, Nov 17, 2017 7:05 PM

Granted I expect someone on this list to reply with something that makes me feel stupid, if you have two GPSDO units running side by side, should the phase delta on the 10Mhz output be zero (ideally)?  Is there an absolute phase standard kept between GPSDO units as all it would take is one extra inverter in the chain to flip the phase, no?

Thanks

Jerry

Granted I expect someone on this list to reply with something that makes me feel stupid, if you have two GPSDO units running side by side, should the phase delta on the 10Mhz output be zero (ideally)? Is there an absolute phase standard kept between GPSDO units as all it would take is one extra inverter in the chain to flip the phase, no? Thanks Jerry
J
jimlux
Fri, Nov 17, 2017 7:15 PM

On 11/17/17 11:05 AM, Jerry Hancock wrote:

Granted I expect someone on this list to reply with something that makes me feel stupid, if you have two GPSDO units running side by side, should the phase delta on the 10Mhz output be zero (ideally)?  Is there an absolute phase standard kept between GPSDO units as all it would take is one extra inverter in the chain to flip the phase, no?

I don't think so..
GPSDOs are really more of a syntonization (same frequency) rather than a
synchronization (same phase) box.

You might have two boxes of the same kind that have some well defined
and stable phase relationship between, say, the 10 MHz sine and the 1pps
(if the 1pps is derived by dividing the 10 MHz, for instance)

On 11/17/17 11:05 AM, Jerry Hancock wrote: > Granted I expect someone on this list to reply with something that makes me feel stupid, if you have two GPSDO units running side by side, should the phase delta on the 10Mhz output be zero (ideally)? Is there an absolute phase standard kept between GPSDO units as all it would take is one extra inverter in the chain to flip the phase, no? > I don't think so.. GPSDOs are really more of a syntonization (same frequency) rather than a synchronization (same phase) box. You might have two boxes of the same kind that have some well defined and stable phase relationship between, say, the 10 MHz sine and the 1pps (if the 1pps is derived by dividing the 10 MHz, for instance)
BK
Bob kb8tq
Fri, Nov 17, 2017 7:23 PM

Hi

Most GPSDO’s use a divider to get straight from the 10 MHz to the PPS. There are some that
derive both the PPS and the 10 MHz off a higher frequency, they are not common in the eBay
world (yet). The divider is going to have a fixed delay based on it’s design. That might be 35 ns,
it might be 3.5 ns. It all depends on how they did this or that. There easily could be an inversion
that moves everything 50 ns.

If the GPSDO wanders around over a 30 ns range, that’s a pretty big phase shift at 10 MHz. If
they both are doing this you are up to (maybe) half a cycle or more.

Now toss in devices that “skip” a 10 MHz cycle to keep the PPS aligned. This is pretty uncommon
in normal operation. Uncommon does not mean that it never happens.

So, if you have two of the same model / same firmware GPSDO’s, the 10 MHz phase is probably
“in the ballpark”. It’s certainly not close in terms of < 1 degree or something like that. If you have
GPSDO’s from multiple outfits …. who knows what you will see.

Bob

On Nov 17, 2017, at 2:05 PM, Jerry Hancock jerry@hanler.com wrote:

Granted I expect someone on this list to reply with something that makes me feel stupid, if you have two GPSDO units running side by side, should the phase delta on the 10Mhz output be zero (ideally)?  Is there an absolute phase standard kept between GPSDO units as all it would take is one extra inverter in the chain to flip the phase, no?

Thanks

Jerry


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Most GPSDO’s use a divider to get straight from the 10 MHz to the PPS. There are some that derive both the PPS and the 10 MHz off a higher frequency, they are not common in the eBay world (yet). The divider is going to have a fixed delay based on it’s design. That might be 35 ns, it might be 3.5 ns. It all depends on how they did this or that. There easily could be an inversion that moves everything 50 ns. If the GPSDO wanders around over a 30 ns range, that’s a pretty big phase shift at 10 MHz. If they both are doing this you are up to (maybe) half a cycle or more. Now toss in devices that “skip” a 10 MHz cycle to keep the PPS aligned. This is pretty uncommon in normal operation. Uncommon does not mean that it never happens. So, if you have two of the same model / same firmware GPSDO’s, the 10 MHz phase is probably “in the ballpark”. It’s certainly not close in terms of < 1 degree or something like that. If you have GPSDO’s from multiple outfits …. who knows what you will see. Bob > On Nov 17, 2017, at 2:05 PM, Jerry Hancock <jerry@hanler.com> wrote: > > Granted I expect someone on this list to reply with something that makes me feel stupid, if you have two GPSDO units running side by side, should the phase delta on the 10Mhz output be zero (ideally)? Is there an absolute phase standard kept between GPSDO units as all it would take is one extra inverter in the chain to flip the phase, no? > > Thanks > > Jerry > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
MD
Magnus Danielson
Fri, Nov 17, 2017 7:39 PM

Hi,

First of all, the receiver varies around some average phase.
The average of two receivers may not be the same, even if sharing antenna.
Their average can be offset from "absolute phase" because of offsets in
the receiver, antenna cables and antenna.

If you are not to careful, a GPSDO will do nicely for many applications.
The more you care, the more hassle.

Oh, and two GPSDOs next to each other can have significant common mode
disturbance, so you can't really evaluate a GPSDO by measuring up to
another GPSDO.

Cheers,
Magnus

On 11/17/2017 08:05 PM, Jerry Hancock wrote:

Granted I expect someone on this list to reply with something that makes me feel stupid, if you have two GPSDO units running side by side, should the phase delta on the 10Mhz output be zero (ideally)?  Is there an absolute phase standard kept between GPSDO units as all it would take is one extra inverter in the chain to flip the phase, no?

Thanks

Jerry


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi, First of all, the receiver varies around some average phase. The average of two receivers may not be the same, even if sharing antenna. Their average can be offset from "absolute phase" because of offsets in the receiver, antenna cables and antenna. If you are not to careful, a GPSDO will do nicely for many applications. The more you care, the more hassle. Oh, and two GPSDOs next to each other can have significant common mode disturbance, so you can't really evaluate a GPSDO by measuring up to another GPSDO. Cheers, Magnus On 11/17/2017 08:05 PM, Jerry Hancock wrote: > Granted I expect someone on this list to reply with something that makes me feel stupid, if you have two GPSDO units running side by side, should the phase delta on the 10Mhz output be zero (ideally)? Is there an absolute phase standard kept between GPSDO units as all it would take is one extra inverter in the chain to flip the phase, no? > > Thanks > > Jerry > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
JH
Jerry Hancock
Fri, Nov 17, 2017 8:01 PM

Thanks, understand completely.

You mention “sharing an antenna”.  Is there a problem with connecting the same antenna to two units?  I am referring to my REF0/REF1 pair.

Thanks

On Nov 17, 2017, at 11:39 AM, Magnus Danielson magnus@rubidium.dyndns.org wrote:

Hi,

First of all, the receiver varies around some average phase.
The average of two receivers may not be the same, even if sharing antenna.
Their average can be offset from "absolute phase" because of offsets in the receiver, antenna cables and antenna.

If you are not to careful, a GPSDO will do nicely for many applications. The more you care, the more hassle.

Oh, and two GPSDOs next to each other can have significant common mode disturbance, so you can't really evaluate a GPSDO by measuring up to another GPSDO.

Cheers,
Magnus

On 11/17/2017 08:05 PM, Jerry Hancock wrote:

Granted I expect someone on this list to reply with something that makes me feel stupid, if you have two GPSDO units running side by side, should the phase delta on the 10Mhz output be zero (ideally)?  Is there an absolute phase standard kept between GPSDO units as all it would take is one extra inverter in the chain to flip the phase, no?
Thanks
Jerry


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Thanks, understand completely. You mention “sharing an antenna”. Is there a problem with connecting the same antenna to two units? I am referring to my REF0/REF1 pair. Thanks > On Nov 17, 2017, at 11:39 AM, Magnus Danielson <magnus@rubidium.dyndns.org> wrote: > > Hi, > > First of all, the receiver varies around some average phase. > The average of two receivers may not be the same, even if sharing antenna. > Their average can be offset from "absolute phase" because of offsets in the receiver, antenna cables and antenna. > > If you are not to careful, a GPSDO will do nicely for many applications. The more you care, the more hassle. > > Oh, and two GPSDOs next to each other can have significant common mode disturbance, so you can't really evaluate a GPSDO by measuring up to another GPSDO. > > Cheers, > Magnus > > On 11/17/2017 08:05 PM, Jerry Hancock wrote: >> Granted I expect someone on this list to reply with something that makes me feel stupid, if you have two GPSDO units running side by side, should the phase delta on the 10Mhz output be zero (ideally)? Is there an absolute phase standard kept between GPSDO units as all it would take is one extra inverter in the chain to flip the phase, no? >> Thanks >> Jerry >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
MD
Magnus Danielson
Fri, Nov 17, 2017 8:20 PM

Hi,

The reason I mention sharing antenna is that it takes out the difference
that is natural from having different antennas and the difference in
multipath.

Sharing antenna has its own set of problems, so a suitable antenna
splitter was assumed. Effectively you need to bypass DC to power the
antenna, you need to amplify the signal to compensate for the loss due
to the splitter, and it needs to look like an antenna to the GPS
receiver, so the antenna alarm does not goes off.

Cheers,
Magnus

On 11/17/2017 09:01 PM, Jerry Hancock wrote:

Thanks, understand completely.

You mention “sharing an antenna”.  Is there a problem with connecting the same antenna to two units?  I am referring to my REF0/REF1 pair.

Thanks

On Nov 17, 2017, at 11:39 AM, Magnus Danielson magnus@rubidium.dyndns.org wrote:

Hi,

First of all, the receiver varies around some average phase.
The average of two receivers may not be the same, even if sharing antenna.
Their average can be offset from "absolute phase" because of offsets in the receiver, antenna cables and antenna.

If you are not to careful, a GPSDO will do nicely for many applications. The more you care, the more hassle.

Oh, and two GPSDOs next to each other can have significant common mode disturbance, so you can't really evaluate a GPSDO by measuring up to another GPSDO.

Cheers,
Magnus

On 11/17/2017 08:05 PM, Jerry Hancock wrote:

Granted I expect someone on this list to reply with something that makes me feel stupid, if you have two GPSDO units running side by side, should the phase delta on the 10Mhz output be zero (ideally)?  Is there an absolute phase standard kept between GPSDO units as all it would take is one extra inverter in the chain to flip the phase, no?
Thanks
Jerry


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi, The reason I mention sharing antenna is that it takes out the difference that is natural from having different antennas and the difference in multipath. Sharing antenna has its own set of problems, so a suitable antenna splitter was assumed. Effectively you need to bypass DC to power the antenna, you need to amplify the signal to compensate for the loss due to the splitter, and it needs to look like an antenna to the GPS receiver, so the antenna alarm does not goes off. Cheers, Magnus On 11/17/2017 09:01 PM, Jerry Hancock wrote: > Thanks, understand completely. > > You mention “sharing an antenna”. Is there a problem with connecting the same antenna to two units? I am referring to my REF0/REF1 pair. > > Thanks > >> On Nov 17, 2017, at 11:39 AM, Magnus Danielson <magnus@rubidium.dyndns.org> wrote: >> >> Hi, >> >> First of all, the receiver varies around some average phase. >> The average of two receivers may not be the same, even if sharing antenna. >> Their average can be offset from "absolute phase" because of offsets in the receiver, antenna cables and antenna. >> >> If you are not to careful, a GPSDO will do nicely for many applications. The more you care, the more hassle. >> >> Oh, and two GPSDOs next to each other can have significant common mode disturbance, so you can't really evaluate a GPSDO by measuring up to another GPSDO. >> >> Cheers, >> Magnus >> >> On 11/17/2017 08:05 PM, Jerry Hancock wrote: >>> Granted I expect someone on this list to reply with something that makes me feel stupid, if you have two GPSDO units running side by side, should the phase delta on the 10Mhz output be zero (ideally)? Is there an absolute phase standard kept between GPSDO units as all it would take is one extra inverter in the chain to flip the phase, no? >>> Thanks >>> Jerry >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >
JH
Jerry Hancock
Fri, Nov 17, 2017 8:25 PM

I figured, what the heck, and just used a BNC T and it’s working.

On Nov 17, 2017, at 12:20 PM, Magnus Danielson magnus@rubidium.se wrote:

Hi,

The reason I mention sharing antenna is that it takes out the difference that is natural from having different antennas and the difference in multipath.

Sharing antenna has its own set of problems, so a suitable antenna splitter was assumed. Effectively you need to bypass DC to power the antenna, you need to amplify the signal to compensate for the loss due to the splitter, and it needs to look like an antenna to the GPS receiver, so the antenna alarm does not goes off.

Cheers,
Magnus

On 11/17/2017 09:01 PM, Jerry Hancock wrote:

Thanks, understand completely.
You mention “sharing an antenna”.  Is there a problem with connecting the same antenna to two units?  I am referring to my REF0/REF1 pair.
Thanks

On Nov 17, 2017, at 11:39 AM, Magnus Danielson magnus@rubidium.dyndns.org wrote:

Hi,

First of all, the receiver varies around some average phase.
The average of two receivers may not be the same, even if sharing antenna.
Their average can be offset from "absolute phase" because of offsets in the receiver, antenna cables and antenna.

If you are not to careful, a GPSDO will do nicely for many applications. The more you care, the more hassle.

Oh, and two GPSDOs next to each other can have significant common mode disturbance, so you can't really evaluate a GPSDO by measuring up to another GPSDO.

Cheers,
Magnus

On 11/17/2017 08:05 PM, Jerry Hancock wrote:

Granted I expect someone on this list to reply with something that makes me feel stupid, if you have two GPSDO units running side by side, should the phase delta on the 10Mhz output be zero (ideally)?  Is there an absolute phase standard kept between GPSDO units as all it would take is one extra inverter in the chain to flip the phase, no?
Thanks
Jerry


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

I figured, what the heck, and just used a BNC T and it’s working. > On Nov 17, 2017, at 12:20 PM, Magnus Danielson <magnus@rubidium.se> wrote: > > Hi, > > The reason I mention sharing antenna is that it takes out the difference that is natural from having different antennas and the difference in multipath. > > Sharing antenna has its own set of problems, so a suitable antenna splitter was assumed. Effectively you need to bypass DC to power the antenna, you need to amplify the signal to compensate for the loss due to the splitter, and it needs to look like an antenna to the GPS receiver, so the antenna alarm does not goes off. > > Cheers, > Magnus > > On 11/17/2017 09:01 PM, Jerry Hancock wrote: >> Thanks, understand completely. >> You mention “sharing an antenna”. Is there a problem with connecting the same antenna to two units? I am referring to my REF0/REF1 pair. >> Thanks >>> On Nov 17, 2017, at 11:39 AM, Magnus Danielson <magnus@rubidium.dyndns.org> wrote: >>> >>> Hi, >>> >>> First of all, the receiver varies around some average phase. >>> The average of two receivers may not be the same, even if sharing antenna. >>> Their average can be offset from "absolute phase" because of offsets in the receiver, antenna cables and antenna. >>> >>> If you are not to careful, a GPSDO will do nicely for many applications. The more you care, the more hassle. >>> >>> Oh, and two GPSDOs next to each other can have significant common mode disturbance, so you can't really evaluate a GPSDO by measuring up to another GPSDO. >>> >>> Cheers, >>> Magnus >>> >>> On 11/17/2017 08:05 PM, Jerry Hancock wrote: >>>> Granted I expect someone on this list to reply with something that makes me feel stupid, if you have two GPSDO units running side by side, should the phase delta on the 10Mhz output be zero (ideally)? Is there an absolute phase standard kept between GPSDO units as all it would take is one extra inverter in the chain to flip the phase, no? >>>> Thanks >>>> Jerry >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>> and follow the instructions there. >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there.
MD
Magnus Danielson
Fri, Nov 17, 2017 8:35 PM

Can work, can go wrong.

One should be careful and think it though.

Cheers,
Magnus

On 11/17/2017 09:25 PM, Jerry Hancock wrote:

I figured, what the heck, and just used a BNC T and it’s working.

On Nov 17, 2017, at 12:20 PM, Magnus Danielson magnus@rubidium.se wrote:

Hi,

The reason I mention sharing antenna is that it takes out the difference that is natural from having different antennas and the difference in multipath.

Sharing antenna has its own set of problems, so a suitable antenna splitter was assumed. Effectively you need to bypass DC to power the antenna, you need to amplify the signal to compensate for the loss due to the splitter, and it needs to look like an antenna to the GPS receiver, so the antenna alarm does not goes off.

Cheers,
Magnus

On 11/17/2017 09:01 PM, Jerry Hancock wrote:

Thanks, understand completely.
You mention “sharing an antenna”.  Is there a problem with connecting the same antenna to two units?  I am referring to my REF0/REF1 pair.
Thanks

On Nov 17, 2017, at 11:39 AM, Magnus Danielson magnus@rubidium.dyndns.org wrote:

Hi,

First of all, the receiver varies around some average phase.
The average of two receivers may not be the same, even if sharing antenna.
Their average can be offset from "absolute phase" because of offsets in the receiver, antenna cables and antenna.

If you are not to careful, a GPSDO will do nicely for many applications. The more you care, the more hassle.

Oh, and two GPSDOs next to each other can have significant common mode disturbance, so you can't really evaluate a GPSDO by measuring up to another GPSDO.

Cheers,
Magnus

On 11/17/2017 08:05 PM, Jerry Hancock wrote:

Granted I expect someone on this list to reply with something that makes me feel stupid, if you have two GPSDO units running side by side, should the phase delta on the 10Mhz output be zero (ideally)?  Is there an absolute phase standard kept between GPSDO units as all it would take is one extra inverter in the chain to flip the phase, no?
Thanks
Jerry


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Can work, can go wrong. One should be careful and think it though. Cheers, Magnus On 11/17/2017 09:25 PM, Jerry Hancock wrote: > I figured, what the heck, and just used a BNC T and it’s working. > > >> On Nov 17, 2017, at 12:20 PM, Magnus Danielson <magnus@rubidium.se> wrote: >> >> Hi, >> >> The reason I mention sharing antenna is that it takes out the difference that is natural from having different antennas and the difference in multipath. >> >> Sharing antenna has its own set of problems, so a suitable antenna splitter was assumed. Effectively you need to bypass DC to power the antenna, you need to amplify the signal to compensate for the loss due to the splitter, and it needs to look like an antenna to the GPS receiver, so the antenna alarm does not goes off. >> >> Cheers, >> Magnus >> >> On 11/17/2017 09:01 PM, Jerry Hancock wrote: >>> Thanks, understand completely. >>> You mention “sharing an antenna”. Is there a problem with connecting the same antenna to two units? I am referring to my REF0/REF1 pair. >>> Thanks >>>> On Nov 17, 2017, at 11:39 AM, Magnus Danielson <magnus@rubidium.dyndns.org> wrote: >>>> >>>> Hi, >>>> >>>> First of all, the receiver varies around some average phase. >>>> The average of two receivers may not be the same, even if sharing antenna. >>>> Their average can be offset from "absolute phase" because of offsets in the receiver, antenna cables and antenna. >>>> >>>> If you are not to careful, a GPSDO will do nicely for many applications. The more you care, the more hassle. >>>> >>>> Oh, and two GPSDOs next to each other can have significant common mode disturbance, so you can't really evaluate a GPSDO by measuring up to another GPSDO. >>>> >>>> Cheers, >>>> Magnus >>>> >>>> On 11/17/2017 08:05 PM, Jerry Hancock wrote: >>>>> Granted I expect someone on this list to reply with something that makes me feel stupid, if you have two GPSDO units running side by side, should the phase delta on the 10Mhz output be zero (ideally)? Is there an absolute phase standard kept between GPSDO units as all it would take is one extra inverter in the chain to flip the phase, no? >>>>> Thanks >>>>> Jerry >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@febo.com >>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>>> and follow the instructions there. >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>> and follow the instructions there. >
JH
Jerry Hancock
Fri, Nov 17, 2017 8:45 PM

I agree.  I disconnected it after wondering about the active antenna bias.

On Nov 17, 2017, at 12:35 PM, Magnus Danielson magnus@rubidium.se wrote:

Can work, can go wrong.

One should be careful and think it though.

Cheers,
Magnus

On 11/17/2017 09:25 PM, Jerry Hancock wrote:

I figured, what the heck, and just used a BNC T and it’s working.

On Nov 17, 2017, at 12:20 PM, Magnus Danielson magnus@rubidium.se wrote:

Hi,

The reason I mention sharing antenna is that it takes out the difference that is natural from having different antennas and the difference in multipath.

Sharing antenna has its own set of problems, so a suitable antenna splitter was assumed. Effectively you need to bypass DC to power the antenna, you need to amplify the signal to compensate for the loss due to the splitter, and it needs to look like an antenna to the GPS receiver, so the antenna alarm does not goes off.

Cheers,
Magnus

On 11/17/2017 09:01 PM, Jerry Hancock wrote:

Thanks, understand completely.
You mention “sharing an antenna”.  Is there a problem with connecting the same antenna to two units?  I am referring to my REF0/REF1 pair.
Thanks

On Nov 17, 2017, at 11:39 AM, Magnus Danielson magnus@rubidium.dyndns.org wrote:

Hi,

First of all, the receiver varies around some average phase.
The average of two receivers may not be the same, even if sharing antenna.
Their average can be offset from "absolute phase" because of offsets in the receiver, antenna cables and antenna.

If you are not to careful, a GPSDO will do nicely for many applications. The more you care, the more hassle.

Oh, and two GPSDOs next to each other can have significant common mode disturbance, so you can't really evaluate a GPSDO by measuring up to another GPSDO.

Cheers,
Magnus

On 11/17/2017 08:05 PM, Jerry Hancock wrote:

Granted I expect someone on this list to reply with something that makes me feel stupid, if you have two GPSDO units running side by side, should the phase delta on the 10Mhz output be zero (ideally)?  Is there an absolute phase standard kept between GPSDO units as all it would take is one extra inverter in the chain to flip the phase, no?
Thanks
Jerry


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

I agree. I disconnected it after wondering about the active antenna bias. > On Nov 17, 2017, at 12:35 PM, Magnus Danielson <magnus@rubidium.se> wrote: > > Can work, can go wrong. > > One should be careful and think it though. > > Cheers, > Magnus > > On 11/17/2017 09:25 PM, Jerry Hancock wrote: >> I figured, what the heck, and just used a BNC T and it’s working. >>> On Nov 17, 2017, at 12:20 PM, Magnus Danielson <magnus@rubidium.se> wrote: >>> >>> Hi, >>> >>> The reason I mention sharing antenna is that it takes out the difference that is natural from having different antennas and the difference in multipath. >>> >>> Sharing antenna has its own set of problems, so a suitable antenna splitter was assumed. Effectively you need to bypass DC to power the antenna, you need to amplify the signal to compensate for the loss due to the splitter, and it needs to look like an antenna to the GPS receiver, so the antenna alarm does not goes off. >>> >>> Cheers, >>> Magnus >>> >>> On 11/17/2017 09:01 PM, Jerry Hancock wrote: >>>> Thanks, understand completely. >>>> You mention “sharing an antenna”. Is there a problem with connecting the same antenna to two units? I am referring to my REF0/REF1 pair. >>>> Thanks >>>>> On Nov 17, 2017, at 11:39 AM, Magnus Danielson <magnus@rubidium.dyndns.org> wrote: >>>>> >>>>> Hi, >>>>> >>>>> First of all, the receiver varies around some average phase. >>>>> The average of two receivers may not be the same, even if sharing antenna. >>>>> Their average can be offset from "absolute phase" because of offsets in the receiver, antenna cables and antenna. >>>>> >>>>> If you are not to careful, a GPSDO will do nicely for many applications. The more you care, the more hassle. >>>>> >>>>> Oh, and two GPSDOs next to each other can have significant common mode disturbance, so you can't really evaluate a GPSDO by measuring up to another GPSDO. >>>>> >>>>> Cheers, >>>>> Magnus >>>>> >>>>> On 11/17/2017 08:05 PM, Jerry Hancock wrote: >>>>>> Granted I expect someone on this list to reply with something that makes me feel stupid, if you have two GPSDO units running side by side, should the phase delta on the 10Mhz output be zero (ideally)? Is there an absolute phase standard kept between GPSDO units as all it would take is one extra inverter in the chain to flip the phase, no? >>>>>> Thanks >>>>>> Jerry >>>>>> _______________________________________________ >>>>>> time-nuts mailing list -- time-nuts@febo.com >>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>>>> and follow the instructions there. >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@febo.com >>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>>> and follow the instructions there.