Some time ago I presented a design where I replaced the non-working chopper
circuitry for a Fluke 887 differential voltmeter with an LTC1050 precision
chopper amplifier. The design seemed to work fine, but I have discovered
an issue ( I haven’t actually used the meter very much). The manual says
to zero the meter by shorting the input terminals and then adjusting the
zero adjustment for a zero reading on the meter. This works fine with the
LTC1050 circuit, but I recently noticed the meter reads -0.08 when the
input is open on the 1V range and the null switch is on 0.0001 (it is fine
on all other ranges/null switch settings). This equates to a -8uV
reading. I believe this is due to the input bias current of the LTC1050
and the 1 megohm input resistance of the 887 on the most sensitive
range/null setting. I tried three different LTC1050s and they all exhibit
the same problem.
However, this would imply an 8pA bias current, which is much higher than
expected. My question is: does anyone else have a better explanation for
this issue and does anyone experience this issue with the original circuit?
Thanks,
Randy Evans AE6YG
Could it be current noise on the LTC1050 inputs? Charge pumping
produces a current noise and bias current spikes which are deceptively
large given the input bias current specification.
On Wed, 26 Apr 2017 19:17:08 -0700, you wrote:
Some time ago I presented a design where I replaced the non-working chopper
circuitry for a Fluke 887 differential voltmeter with an LTC1050 precision
chopper amplifier. The design seemed to work fine, but I have discovered
an issue ( I havent actually used the meter very much). The manual says
to zero the meter by shorting the input terminals and then adjusting the
zero adjustment for a zero reading on the meter. This works fine with the
LTC1050 circuit, but I recently noticed the meter reads -0.08 when the
input is open on the 1V range and the null switch is on 0.0001 (it is fine
on all other ranges/null switch settings). This equates to a -8uV
reading. I believe this is due to the input bias current of the LTC1050
and the 1 megohm input resistance of the 887 on the most sensitive
range/null setting. I tried three different LTC1050s and they all exhibit
the same problem.
However, this would imply an 8pA bias current, which is much higher than
expected. My question is: does anyone else have a better explanation for
this issue and does anyone experience this issue with the original circuit?
Thanks,
Randy Evans AE6YG
Randy wrote:
but I recently noticed the meter reads -0.08 when the
input is open on the 1V range and the null switch is on 0.0001
This equates to a -8uV reading.
I believe this is due to the input bias current of the LTC1050
* * *
However, this would imply an 8pA bias current,
which is much higher than expected.
I am attaching the relevant portion of the LTC1050 datasheet. Note that
at 25C, the 1050 has an input offset current spec of +/- 20pA (both
typical -- max is ~5x worse). I would expect a total input offset of
+/- 20uV (typical) in your circuit. (The 0.5uV typical input offset
voltage is not a significant contributor in this context.)
Best regards
Charles
Charles,
You are absolutely correct. I misread the data sheet. It's interesting
that all three LTC1050s behave exactly the same. They have the same lot
code number.but I would expect more variation. Live an learn.
Randy
On Thu, Apr 27, 2017 at 12:18 PM, Charles Steinmetz csteinmetz@yandex.com
wrote:
Randy wrote:
but I recently noticed the meter reads -0.08 when the
input is open on the 1V range and the null switch is on 0.0001
This equates to a -8uV reading.
I believe this is due to the input bias current of the LTC1050
* * *
However, this would imply an 8pA bias current,
which is much higher than expected.
I am attaching the relevant portion of the LTC1050 datasheet. Note that
at 25C, the 1050 has an input offset current spec of +/- 20pA (both typical
-- max is ~5x worse). I would expect a total input offset of +/- 20uV
(typical) in your circuit. (The 0.5uV typical input offset voltage is not
a significant contributor in this context.)
Best regards
Charles
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
It looks like a better candidate would be the LTC2054, which has a typical
bias current of 1 pA at 25C and is essentially constant up to 45C.
Randy
On Thu, Apr 27, 2017 at 5:02 PM, Randy Evans randyevans2688@gmail.com
wrote:
Charles,
You are absolutely correct. I misread the data sheet. It's interesting
that all three LTC1050s behave exactly the same. They have the same lot
code number.but I would expect more variation. Live an learn.
Randy
On Thu, Apr 27, 2017 at 12:18 PM, Charles Steinmetz <csteinmetz@yandex.com
wrote:
Randy wrote:
but I recently noticed the meter reads -0.08 when the
input is open on the 1V range and the null switch is on 0.0001
This equates to a -8uV reading.
I believe this is due to the input bias current of the LTC1050
* * *
However, this would imply an 8pA bias current,
which is much higher than expected.
I am attaching the relevant portion of the LTC1050 datasheet. Note that
at 25C, the 1050 has an input offset current spec of +/- 20pA (both typical
-- max is ~5x worse). I would expect a total input offset of +/- 20uV
(typical) in your circuit. (The 0.5uV typical input offset voltage is not
a significant contributor in this context.)
Best regards
Charles