volt-nuts@lists.febo.com

Discussion of precise voltage measurement

View all threads

Service Manual vintage DMM Fluke 7600A

CD
cgm de bruijn
Mon, Nov 5, 2018 9:00 PM

Hi, all

I have not been able to find a service and/or user manual for a vintage
(1978-1982) DMM Fluke 7600A that I picked up and want to check/calibrate
against my calibrated Keithley 2001 multimeter.

The instrument is still in good shape and up and running:  offset and gain
need adjustment, but how?

Does anyone know of a copy of the user manual/schematic?  A helpful Fluke
technician came up with the 8502A manual that comes close, but gives no
calibration procedure. Further surfing on Google had no success.

Thanks in advance,

Kees de Bruijn

cgmdebruijn@gmail.com

Hi, all I have not been able to find a service and/or user manual for a vintage (1978-1982) DMM Fluke 7600A that I picked up and want to check/calibrate against my calibrated Keithley 2001 multimeter. The instrument is still in good shape and up and running: offset and gain need adjustment, but how? Does anyone know of a copy of the user manual/schematic? A helpful Fluke technician came up with the 8502A manual that comes close, but gives no calibration procedure. Further surfing on Google had no success. Thanks in advance, Kees de Bruijn cgmdebruijn@gmail.com
RK
Rob Klein
Tue, Nov 6, 2018 3:26 PM

Hi Kees,

The 8502A manual on KO4BB's site does contain a section on calibration
adjustments.
http://www.ko4bb.com/getsimple/index.php?id=download&file=Fluke/Fluke_8502A_6_12_Digit_Multimeter_Service_Manual.pdf

Hope that helps you a bit.

Met vriendelijke groet,
Rob.

Op 5-11-2018 om 22:00 schreef cgm de bruijn:

Hi, all

I have not been able to find a service and/or user manual for a vintage
(1978-1982) DMM Fluke 7600A that I picked up and want to check/calibrate
against my calibrated Keithley 2001 multimeter.

The instrument is still in good shape and up and running:  offset and gain
need adjustment, but how?

Does anyone know of a copy of the user manual/schematic?  A helpful Fluke
technician came up with the 8502A manual that comes close, but gives no
calibration procedure. Further surfing on Google had no success.

Thanks in advance,

Kees de Bruijn

cgmdebruijn@gmail.com


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.

Hi Kees, The 8502A manual on KO4BB's site does contain a section on calibration adjustments. http://www.ko4bb.com/getsimple/index.php?id=download&file=Fluke/Fluke_8502A_6_12_Digit_Multimeter_Service_Manual.pdf Hope that helps you a bit. Met vriendelijke groet, Rob. Op 5-11-2018 om 22:00 schreef cgm de bruijn: > Hi, all > > I have not been able to find a service and/or user manual for a vintage > (1978-1982) DMM Fluke 7600A that I picked up and want to check/calibrate > against my calibrated Keithley 2001 multimeter. > > The instrument is still in good shape and up and running: offset and gain > need adjustment, but how? > > Does anyone know of a copy of the user manual/schematic? A helpful Fluke > technician came up with the 8502A manual that comes close, but gives no > calibration procedure. Further surfing on Google had no success. > > Thanks in advance, > > > > > Kees de Bruijn > > cgmdebruijn@gmail.com > _______________________________________________ > volt-nuts mailing list -- volt-nuts@lists.febo.com > To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com > and follow the instructions there. >
RP
Roy Phillips
Tue, Nov 6, 2018 4:04 PM

Kees
You perhaps should try the Fluke (Australia) website, as in the past they
offered more manuals than any other Fluke site.
Roy

-----Original Message-----
From: cgm de bruijn
Sent: Monday, November 05, 2018 9:00 PM
To: volt-nuts@lists.febo.com
Subject: [volt-nuts] Service Manual vintage DMM Fluke 7600A

Hi, all

I have not been able to find a service and/or user manual for a vintage
(1978-1982) DMM Fluke 7600A that I picked up and want to check/calibrate
against my calibrated Keithley 2001 multimeter.

The instrument is still in good shape and up and running:  offset and gain
need adjustment, but how?

Does anyone know of a copy of the user manual/schematic?  A helpful Fluke
technician came up with the 8502A manual that comes close, but gives no
calibration procedure. Further surfing on Google had no success.

Thanks in advance,

Kees de Bruijn

cgmdebruijn@gmail.com


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.

Kees You perhaps should try the Fluke (Australia) website, as in the past they offered more manuals than any other Fluke site. Roy -----Original Message----- From: cgm de bruijn Sent: Monday, November 05, 2018 9:00 PM To: volt-nuts@lists.febo.com Subject: [volt-nuts] Service Manual vintage DMM Fluke 7600A Hi, all I have not been able to find a service and/or user manual for a vintage (1978-1982) DMM Fluke 7600A that I picked up and want to check/calibrate against my calibrated Keithley 2001 multimeter. The instrument is still in good shape and up and running: offset and gain need adjustment, but how? Does anyone know of a copy of the user manual/schematic? A helpful Fluke technician came up with the 8502A manual that comes close, but gives no calibration procedure. Further surfing on Google had no success. Thanks in advance, Kees de Bruijn cgmdebruijn@gmail.com _______________________________________________ volt-nuts mailing list -- volt-nuts@lists.febo.com To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com and follow the instructions there.