I am interested to hear people's view on the relative merits of the
HP/Agilent/ Keysight 3458A vs the Keithley 2002.
I noticed that when I recently visit the standard lab for voltage at NPL,
the place was littered with Agilent 3458As, but I did not notice a single
Keithley 2002. But maybe NPL get better discounts from HP/Agilent Keysight
than Keithley/Tektronix.
I notice that the used prices of 3458A is higher than 2002s.
Also, are there any things ons should be looking out for when buying a
3458A or 2002? I have in the past thought an Agilent front panel was a
better than an HP one, as it would not be as old, but this is not always
true - see my comment later.
What S/N are best avoided? Does anyone have a record of S/Ns of Keighley
meters with time?
Very recently I had a somewhat scruffy HP 6674A PSU with a damaged LCD. It
cost me $50 or so to pick up a clean Agilent front panel with a new
display. The display actually turned out to be unsuitable, yet my Agilent
badged 6674A is actually a fairly old HP unit. So white my intention was
not to rebadge the PSU, that happened as a result of a legitimate reason to
get it working. So this really indicates how an old HP 3457A could be made
to look a newer Agilent one.
Dave.
Well, it's the topic many can go on and on about.
As with everything there are specific cons and pros of both units, but
here are my 5c:
Now regarding industry standard 3458A. Reason why you saw only 3458A in
metrology lab is simple:
Superior ADC (many rightfully think it's best on market, ever), amazing
linearity (JJA was required to test linearity with good confidence),
very easy calibration (just need 10V reference and 10KOhm standard
resistor to get 3458A calibrated to uncertainty of used standards),
Artifact calibration (ACAL) to significantly remove impact of
temperature variation on accuracy, proven performance. Many labs use
3458A as main transfer standard due to it's ADC performance.
It comes at cost though, as key components of meter's A/D are custom
hybrids. Meaning if you got sour pill, and A/D drifting, you will have
to replace expensive A/D board. Newer Agilent/Keysight branded 3458A's
are essentially same boxes, just with newer hardware and likely with new
FPGA-based digital brains. Performance is same though, unless you really
getting early 1990's meter. Many design issues and items I got covered
in my repair worklog: https://xdevs.com/fix/hp3458a/
Brief 10V comparison : https://xdevs.com/datalog_tcomp/ - sampling 10V
connected to 2001+2002+3458A
Some of the A/D noise comparisons (meter input shorted) -
https://xdevs.com/article/dmm_noise/
So decide what fit best your needs. 3458A is overall higher performance
instrument, but in lot of applications outside of metrology it's
performance benefits may not cover up for higher price and size. Both
3458A and 2002 are still supported and sold new by manufacturers.
Regular calibration of these is likely be over 1000$USD, and lab-grade
calibration for 3458A is usually in range of 2600$USD.
P.S. there are also few other 8.5d meters, but much less information
available on those, as less people have them. Only "modern" 8.5d meter
is 8508A, which cost even more than 3458A without much performance
improvement.
18/07/16 17:54, Dr. David Kirkby (Kirkby Microwave Ltd) пишет:
I am interested to hear people's view on the relative merits of the
HP/Agilent/ Keysight 3458A vs the Keithley 2002.
I noticed that when I recently visit the standard lab for voltage at NPL,
the place was littered with Agilent 3458As, but I did not notice a single
Keithley 2002. But maybe NPL get better discounts from HP/Agilent Keysight
than Keithley/Tektronix.
I notice that the used prices of 3458A is higher than 2002s.
Also, are there any things ons should be looking out for when buying a
3458A or 2002? I have in the past thought an Agilent front panel was a
better than an HP one, as it would not be as old, but this is not always
true - see my comment later.
What S/N are best avoided? Does anyone have a record of S/Ns of Keighley
meters with time?
Very recently I had a somewhat scruffy HP 6674A PSU with a damaged LCD. It
cost me $50 or so to pick up a clean Agilent front panel with a new
display. The display actually turned out to be unsuitable, yet my Agilent
badged 6674A is actually a fairly old HP unit. So white my intention was
not to rebadge the PSU, that happened as a result of a legitimate reason to
get it working. So this really indicates how an old HP 3457A could be made
to look a newer Agilent one.
Dave.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Most labs never turn 3458As off. The longer they are on the lower the drift
rate. I would rather have an 20 year old meter with new batteries than a
new meter.
On Mon, Jul 18, 2016 at 4:11 AM, Illya Tsemenko illya@xdevs.com wrote:
Well, it's the topic many can go on and on about.
As with everything there are specific cons and pros of both units, but
here are my 5c:
Now regarding industry standard 3458A. Reason why you saw only 3458A in
metrology lab is simple:
Superior ADC (many rightfully think it's best on market, ever), amazing
linearity (JJA was required to test linearity with good confidence), very
easy calibration (just need 10V reference and 10KOhm standard resistor to
get 3458A calibrated to uncertainty of used standards), Artifact
calibration (ACAL) to significantly remove impact of temperature variation
on accuracy, proven performance. Many labs use 3458A as main transfer
standard due to it's ADC performance.
It comes at cost though, as key components of meter's A/D are custom
hybrids. Meaning if you got sour pill, and A/D drifting, you will have to
replace expensive A/D board. Newer Agilent/Keysight branded 3458A's are
essentially same boxes, just with newer hardware and likely with new
FPGA-based digital brains. Performance is same though, unless you really
getting early 1990's meter. Many design issues and items I got covered in
my repair worklog: https://xdevs.com/fix/hp3458a/
Brief 10V comparison : https://xdevs.com/datalog_tcomp/ - sampling 10V
connected to 2001+2002+3458A
Some of the A/D noise comparisons (meter input shorted) -
https://xdevs.com/article/dmm_noise/
So decide what fit best your needs. 3458A is overall higher performance
instrument, but in lot of applications outside of metrology it's
performance benefits may not cover up for higher price and size. Both 3458A
and 2002 are still supported and sold new by manufacturers. Regular
calibration of these is likely be over 1000$USD, and lab-grade calibration
for 3458A is usually in range of 2600$USD.
P.S. there are also few other 8.5d meters, but much less information
available on those, as less people have them. Only "modern" 8.5d meter is
8508A, which cost even more than 3458A without much performance improvement.
18/07/16 17:54, Dr. David Kirkby (Kirkby Microwave Ltd) пишет:
I am interested to hear people's view on the relative merits of the
HP/Agilent/ Keysight 3458A vs the Keithley 2002.
I noticed that when I recently visit the standard lab for voltage at NPL,
the place was littered with Agilent 3458As, but I did not notice a single
Keithley 2002. But maybe NPL get better discounts from HP/Agilent Keysight
than Keithley/Tektronix.
I notice that the used prices of 3458A is higher than 2002s.
Also, are there any things ons should be looking out for when buying a
3458A or 2002? I have in the past thought an Agilent front panel was a
better than an HP one, as it would not be as old, but this is not always
true - see my comment later.
What S/N are best avoided? Does anyone have a record of S/Ns of Keighley
meters with time?
Very recently I had a somewhat scruffy HP 6674A PSU with a damaged LCD. It
cost me $50 or so to pick up a clean Agilent front panel with a new
display. The display actually turned out to be unsuitable, yet my
Agilent
badged 6674A is actually a fairly old HP unit. So white my intention was
not to rebadge the PSU, that happened as a result of a legitimate reason
to
get it working. So this really indicates how an old HP 3457A could be made
to look a newer Agilent one.
Dave.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
--
John Phillips
On 18 Jul 2016 17:34, "John Phillips" john.phillips0@gmail.com wrote:
Most labs never turn 3458As off. The longer they are on the lower the
drift
rate. I would rather have an 20 year old meter with new batteries than a
new meter.
If one buys a used meter, one has no idea if it has been switched on very
occasionally or on 24/7.
Do you keep your meter powered up during the transit between your lab and
the cal lab? I would think that quite difficult if you used a courier, but
not a problem if it was taken in a car.
Dave.
We do not keep them powered up for transit.
All the meter gets more stable with age. The reference gets more stable
with burn in.
We store 3458As energised even when we do not plan to use them for a while.
Most 10 year old meters are as stable as the new meters with the high
stability option.
I do not have any stats but quite often we would buy an old meter that had
a old cal sticker and find that 10 volse was still in spec.
On Mon, Jul 18, 2016 at 11:48 AM, Dr. David Kirkby (Kirkby Microwave Ltd) <
drkirkby@kirkbymicrowave.co.uk> wrote:
On 18 Jul 2016 17:34, "John Phillips" john.phillips0@gmail.com wrote:
Most labs never turn 3458As off. The longer they are on the lower the
drift
rate. I would rather have an 20 year old meter with new batteries than a
new meter.
If one buys a used meter, one has no idea if it has been switched on very
occasionally or on 24/7.
Do you keep your meter powered up during the transit between your lab and
the cal lab? I would think that quite difficult if you used a courier, but
not a problem if it was taken in a car.