Why discipline Rubidium oscillator?

JH
Jerry Hancock
Mon, Nov 20, 2017 6:34 PM

I know this is going to sound dumb as I know many GPSDOs had rubidium oscillators in them.  I can see why, in that during holdover, they would tend to be more stable vs others, but given that there is a direct mathematical relationship between the rubidium frequency and potentially the 10Mhz desired output frequency, why do they have to be disciplined or better yet, what advantage does it bring?  Also, I can see how you discipline a DOCXO with the external voltage, how do you discipline a rubidium?  Pulse stretching?

I guess I don’t understand how the technology works, but it seems like an RF signal is swept that would be used to detect a dip at a pretty well defined frequency.  This dip can be used to discipline the oscillator to something like 9Ghz or a factor of what, 900+ times better than 10Mhz.  So wouldn’t that be able to get your desired 10Mhz to 10,000,000.001 or pretty much my level of measurement?  Or does is the dip not quite that precise?  If you can point me to a write-up on this I’ll go away.

Thanks to Gilbert for providing me with at least one rubidium oscillator that is working out of 5 though 2 others seems to stay locked for a few hours during my testing.

Jerry

I know this is going to sound dumb as I know many GPSDOs had rubidium oscillators in them. I can see why, in that during holdover, they would tend to be more stable vs others, but given that there is a direct mathematical relationship between the rubidium frequency and potentially the 10Mhz desired output frequency, why do they have to be disciplined or better yet, what advantage does it bring? Also, I can see how you discipline a DOCXO with the external voltage, how do you discipline a rubidium? Pulse stretching? I guess I don’t understand how the technology works, but it seems like an RF signal is swept that would be used to detect a dip at a pretty well defined frequency. This dip can be used to discipline the oscillator to something like 9Ghz or a factor of what, 900+ times better than 10Mhz. So wouldn’t that be able to get your desired 10Mhz to 10,000,000.001 or pretty much my level of measurement? Or does is the dip not quite that precise? If you can point me to a write-up on this I’ll go away. Thanks to Gilbert for providing me with at least one rubidium oscillator that is working out of 5 though 2 others seems to stay locked for a few hours during my testing. Jerry
BK
Bob kb8tq
Mon, Nov 20, 2017 7:40 PM

Hi

There is no direct relation for an Rb to 10 MYz. Cs beam tubes are what have a direct relation.
Even then, the qualifier is “under standard conditions”. They are sensitive to magnetic field. Rb’s
also are sensitive to magnetic field. Both can be tuned by varying the field. In the case of an Rb
that also takes care of a multitude of other issues.

In the case of Rb, there is a distribution of cells coming out of the manufacturing process. Some
are pretty close to the “right” frequency. Others are way off (as in 100’s of KHz or more). All of them
are capable of meeting the required specs. DDS techniques allow those cells to be used in a
production part. That increases the yield and thus drops the production cost.

Since you now magically have a DDS in the Rb, you can do all sorts of interesting things. If you
suddenly need a 9.99900 MHz standard …. here it is … If you need to do temperature compensation
via a lookup table … it just takes a bit of testing and some code to make it happen. Indeed, the DDS
does also give you some issues. Without some sort of cleanup oscillator, you will have spurs and
phase noise on the output.

Lots of fun ….

Bob

On Nov 20, 2017, at 1:34 PM, Jerry Hancock jerry@hanler.com wrote:

I know this is going to sound dumb as I know many GPSDOs had rubidium oscillators in them.  I can see why, in that during holdover, they would tend to be more stable vs others, but given that there is a direct mathematical relationship between the rubidium frequency and potentially the 10Mhz desired output frequency, why do they have to be disciplined or better yet, what advantage does it bring?  Also, I can see how you discipline a DOCXO with the external voltage, how do you discipline a rubidium?  Pulse stretching?

I guess I don’t understand how the technology works, but it seems like an RF signal is swept that would be used to detect a dip at a pretty well defined frequency.  This dip can be used to discipline the oscillator to something like 9Ghz or a factor of what, 900+ times better than 10Mhz.  So wouldn’t that be able to get your desired 10Mhz to 10,000,000.001 or pretty much my level of measurement?  Or does is the dip not quite that precise?  If you can point me to a write-up on this I’ll go away.

Thanks to Gilbert for providing me with at least one rubidium oscillator that is working out of 5 though 2 others seems to stay locked for a few hours during my testing.

Jerry


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi There is no direct relation for an Rb to 10 MYz. Cs beam tubes are what have a direct relation. Even then, the qualifier is “under standard conditions”. They are sensitive to magnetic field. Rb’s also are sensitive to magnetic field. Both can be tuned by varying the field. In the case of an Rb that also takes care of a multitude of other issues. In the case of Rb, there is a distribution of cells coming out of the manufacturing process. Some are pretty close to the “right” frequency. Others are way off (as in 100’s of KHz or more). All of them are capable of meeting the required specs. DDS techniques allow those cells to be used in a production part. That increases the yield and thus drops the production cost. Since you now magically have a DDS in the Rb, you can do all sorts of interesting things. If you suddenly need a 9.99900 MHz standard …. here it is … If you need to do temperature compensation via a lookup table … it just takes a bit of testing and some code to make it happen. Indeed, the DDS does also give you some issues. Without some sort of cleanup oscillator, you will have spurs and phase noise on the output. Lots of fun …. Bob > On Nov 20, 2017, at 1:34 PM, Jerry Hancock <jerry@hanler.com> wrote: > > I know this is going to sound dumb as I know many GPSDOs had rubidium oscillators in them. I can see why, in that during holdover, they would tend to be more stable vs others, but given that there is a direct mathematical relationship between the rubidium frequency and potentially the 10Mhz desired output frequency, why do they have to be disciplined or better yet, what advantage does it bring? Also, I can see how you discipline a DOCXO with the external voltage, how do you discipline a rubidium? Pulse stretching? > > I guess I don’t understand how the technology works, but it seems like an RF signal is swept that would be used to detect a dip at a pretty well defined frequency. This dip can be used to discipline the oscillator to something like 9Ghz or a factor of what, 900+ times better than 10Mhz. So wouldn’t that be able to get your desired 10Mhz to 10,000,000.001 or pretty much my level of measurement? Or does is the dip not quite that precise? If you can point me to a write-up on this I’ll go away. > > Thanks to Gilbert for providing me with at least one rubidium oscillator that is working out of 5 though 2 others seems to stay locked for a few hours during my testing. > > Jerry > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
DW
Dana Whitlow
Mon, Nov 20, 2017 7:53 PM

And even without problems like external magnetic fields, Rb oscillators do
drift with
age. Over a period of several years they may drift as much as ~1E-9, which
is a huge
error for serious time nuts.

In my pre-retirement job I rode herd on an active Hydrogen maser system,
and even
that has a clear drift tendency.  Generally a couple or three times per
year I had to
make a frequency adjustment in the neighborhood of 3E-14.  And still being
privy to
its performance, I was amused to note that its drift tendency was
interrupted by the
hurricane Maria.  On the day of eye passage over the site the frequency
suddenly
decreased by a few parts in 10^14, held about constant for roughly a week,
then
resumed almost its original value and drift rate thereafter.  If anybody in
this group
can explain* that* behavior (that is, held for a week before resuming old
habits), I'd
love to learn about it.

Dana

On Mon, Nov 20, 2017 at 1:40 PM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

There is no direct relation for an Rb to 10 MYz. Cs beam tubes are what
have a direct relation.
Even then, the qualifier is “under standard conditions”. They are
sensitive to magnetic field. Rb’s
also are sensitive to magnetic field. Both can be tuned by varying the
field. In the case of an Rb
that also takes care of a multitude of other issues.

In the case of Rb, there is a distribution of cells coming out of the
manufacturing process. Some
are pretty close to the “right” frequency. Others are way off (as in 100’s
of KHz or more). All of them
are capable of meeting the required specs. DDS techniques allow those
cells to be used in a
production part. That increases the yield and thus drops the production
cost.

Since you now magically have a DDS in the Rb, you can do all sorts of
interesting things. If you
suddenly need a 9.99900 MHz standard …. here it is … If you need to do
temperature compensation
via a lookup table … it just takes a bit of testing and some code to make
it happen. Indeed, the DDS
does also give you some issues. Without some sort of cleanup oscillator,
you will have spurs and
phase noise on the output.

Lots of fun ….

Bob

On Nov 20, 2017, at 1:34 PM, Jerry Hancock jerry@hanler.com wrote:

I know this is going to sound dumb as I know many GPSDOs had rubidium

oscillators in them.  I can see why, in that during holdover, they would
tend to be more stable vs others, but given that there is a direct
mathematical relationship between the rubidium frequency and potentially
the 10Mhz desired output frequency, why do they have to be disciplined or
better yet, what advantage does it bring?  Also, I can see how you
discipline a DOCXO with the external voltage, how do you discipline a
rubidium?  Pulse stretching?

I guess I don’t understand how the technology works, but it seems like

an RF signal is swept that would be used to detect a dip at a pretty well
defined frequency.  This dip can be used to discipline the oscillator to
something like 9Ghz or a factor of what, 900+ times better than 10Mhz.  So
wouldn’t that be able to get your desired 10Mhz to 10,000,000.001 or pretty
much my level of measurement?  Or does is the dip not quite that precise?
If you can point me to a write-up on this I’ll go away.

Thanks to Gilbert for providing me with at least one rubidium oscillator

that is working out of 5 though 2 others seems to stay locked for a few
hours during my testing.

Jerry


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

And even without problems like external magnetic fields, Rb oscillators do drift with age. Over a period of several years they may drift as much as ~1E-9, which is a *huge* error for serious time nuts. In my pre-retirement job I rode herd on an active Hydrogen maser system, and even that has a clear drift tendency. Generally a couple or three times per year I had to make a frequency adjustment in the neighborhood of 3E-14. And still being privy to its performance, I was amused to note that its drift tendency was interrupted by the hurricane Maria. On the day of eye passage over the site the frequency suddenly decreased by a few parts in 10^14, held about constant for roughly a week, then resumed almost its original value and drift rate thereafter. If anybody in this group can explain* that* behavior (that is, held for a week before resuming old habits), I'd love to learn about it. Dana On Mon, Nov 20, 2017 at 1:40 PM, Bob kb8tq <kb8tq@n1k.org> wrote: > Hi > > There is no direct relation for an Rb to 10 MYz. Cs beam tubes are what > have a direct relation. > Even then, the qualifier is “under standard conditions”. They are > sensitive to magnetic field. Rb’s > also are sensitive to magnetic field. Both can be tuned by varying the > field. In the case of an Rb > that also takes care of a multitude of other issues. > > In the case of Rb, there is a distribution of cells coming out of the > manufacturing process. Some > are pretty close to the “right” frequency. Others are way off (as in 100’s > of KHz or more). All of them > are capable of meeting the required specs. DDS techniques allow those > cells to be used in a > production part. That increases the yield and thus drops the production > cost. > > Since you now magically have a DDS in the Rb, you can do all sorts of > interesting things. If you > suddenly need a 9.99900 MHz standard …. here it is … If you need to do > temperature compensation > via a lookup table … it just takes a bit of testing and some code to make > it happen. Indeed, the DDS > does also give you some issues. Without some sort of cleanup oscillator, > you will have spurs and > phase noise on the output. > > Lots of fun …. > > Bob > > > > On Nov 20, 2017, at 1:34 PM, Jerry Hancock <jerry@hanler.com> wrote: > > > > I know this is going to sound dumb as I know many GPSDOs had rubidium > oscillators in them. I can see why, in that during holdover, they would > tend to be more stable vs others, but given that there is a direct > mathematical relationship between the rubidium frequency and potentially > the 10Mhz desired output frequency, why do they have to be disciplined or > better yet, what advantage does it bring? Also, I can see how you > discipline a DOCXO with the external voltage, how do you discipline a > rubidium? Pulse stretching? > > > > I guess I don’t understand how the technology works, but it seems like > an RF signal is swept that would be used to detect a dip at a pretty well > defined frequency. This dip can be used to discipline the oscillator to > something like 9Ghz or a factor of what, 900+ times better than 10Mhz. So > wouldn’t that be able to get your desired 10Mhz to 10,000,000.001 or pretty > much my level of measurement? Or does is the dip not quite that precise? > If you can point me to a write-up on this I’ll go away. > > > > Thanks to Gilbert for providing me with at least one rubidium oscillator > that is working out of 5 though 2 others seems to stay locked for a few > hours during my testing. > > > > Jerry > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > and follow the instructions there. >
JH
Jerry Hancock
Mon, Nov 20, 2017 8:28 PM

Bob, I was referring to the rubidium standard of 6834682610.904 Hz.  For some reason I thought it was closer to 9Ghz.

I assume then rubidium standards oscillate (if that is the correct term) somewhere around that number but not exact or is it in the detection where things fall down?

On Nov 20, 2017, at 11:40 AM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

There is no direct relation for an Rb to 10 MYz. Cs beam tubes are what have a direct relation.
Even then, the qualifier is “under standard conditions”. They are sensitive to magnetic field. Rb’s
also are sensitive to magnetic field. Both can be tuned by varying the field. In the case of an Rb
that also takes care of a multitude of other issues.

In the case of Rb, there is a distribution of cells coming out of the manufacturing process. Some
are pretty close to the “right” frequency. Others are way off (as in 100’s of KHz or more). All of them
are capable of meeting the required specs. DDS techniques allow those cells to be used in a
production part. That increases the yield and thus drops the production cost.

Since you now magically have a DDS in the Rb, you can do all sorts of interesting things. If you
suddenly need a 9.99900 MHz standard …. here it is … If you need to do temperature compensation
via a lookup table … it just takes a bit of testing and some code to make it happen. Indeed, the DDS
does also give you some issues. Without some sort of cleanup oscillator, you will have spurs and
phase noise on the output.

Lots of fun ….

Bob

On Nov 20, 2017, at 1:34 PM, Jerry Hancock jerry@hanler.com wrote:

I know this is going to sound dumb as I know many GPSDOs had rubidium oscillators in them.  I can see why, in that during holdover, they would tend to be more stable vs others, but given that there is a direct mathematical relationship between the rubidium frequency and potentially the 10Mhz desired output frequency, why do they have to be disciplined or better yet, what advantage does it bring?  Also, I can see how you discipline a DOCXO with the external voltage, how do you discipline a rubidium?  Pulse stretching?

I guess I don’t understand how the technology works, but it seems like an RF signal is swept that would be used to detect a dip at a pretty well defined frequency.  This dip can be used to discipline the oscillator to something like 9Ghz or a factor of what, 900+ times better than 10Mhz.  So wouldn’t that be able to get your desired 10Mhz to 10,000,000.001 or pretty much my level of measurement?  Or does is the dip not quite that precise?  If you can point me to a write-up on this I’ll go away.

Thanks to Gilbert for providing me with at least one rubidium oscillator that is working out of 5 though 2 others seems to stay locked for a few hours during my testing.

Jerry


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Bob, I was referring to the rubidium standard of 6834682610.904 Hz. For some reason I thought it was closer to 9Ghz. I assume then rubidium standards oscillate (if that is the correct term) somewhere around that number but not exact or is it in the detection where things fall down? > On Nov 20, 2017, at 11:40 AM, Bob kb8tq <kb8tq@n1k.org> wrote: > > Hi > > There is no direct relation for an Rb to 10 MYz. Cs beam tubes are what have a direct relation. > Even then, the qualifier is “under standard conditions”. They are sensitive to magnetic field. Rb’s > also are sensitive to magnetic field. Both can be tuned by varying the field. In the case of an Rb > that also takes care of a multitude of other issues. > > In the case of Rb, there is a distribution of cells coming out of the manufacturing process. Some > are pretty close to the “right” frequency. Others are way off (as in 100’s of KHz or more). All of them > are capable of meeting the required specs. DDS techniques allow those cells to be used in a > production part. That increases the yield and thus drops the production cost. > > Since you now magically have a DDS in the Rb, you can do all sorts of interesting things. If you > suddenly need a 9.99900 MHz standard …. here it is … If you need to do temperature compensation > via a lookup table … it just takes a bit of testing and some code to make it happen. Indeed, the DDS > does also give you some issues. Without some sort of cleanup oscillator, you will have spurs and > phase noise on the output. > > Lots of fun …. > > Bob > > >> On Nov 20, 2017, at 1:34 PM, Jerry Hancock <jerry@hanler.com> wrote: >> >> I know this is going to sound dumb as I know many GPSDOs had rubidium oscillators in them. I can see why, in that during holdover, they would tend to be more stable vs others, but given that there is a direct mathematical relationship between the rubidium frequency and potentially the 10Mhz desired output frequency, why do they have to be disciplined or better yet, what advantage does it bring? Also, I can see how you discipline a DOCXO with the external voltage, how do you discipline a rubidium? Pulse stretching? >> >> I guess I don’t understand how the technology works, but it seems like an RF signal is swept that would be used to detect a dip at a pretty well defined frequency. This dip can be used to discipline the oscillator to something like 9Ghz or a factor of what, 900+ times better than 10Mhz. So wouldn’t that be able to get your desired 10Mhz to 10,000,000.001 or pretty much my level of measurement? Or does is the dip not quite that precise? If you can point me to a write-up on this I’ll go away. >> >> Thanks to Gilbert for providing me with at least one rubidium oscillator that is working out of 5 though 2 others seems to stay locked for a few hours during my testing. >> >> Jerry >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
GM
Gregory Maxwell
Mon, Nov 20, 2017 8:39 PM

On Mon, Nov 20, 2017 at 8:28 PM, Jerry Hancock jerry@hanler.com wrote:

Bob, I was referring to the rubidium standard of 6834682610.904 Hz.  For some reason I thought it was closer to 9Ghz.

I assume then rubidium standards oscillate (if that is the correct term) somewhere around that number but not exact or is it in the detection where things fall down?

I think you are confused by the difference between primary and
secondary standards.

A typical rb gas cell is a secondary standard.  Its exact frequency is
distorted by a number of factors like gas pressure, interaction with
the cell walls, and ambient magnetic fields which cannot be canceled
by the design of the standard.  This is why it is useful to discipline
a telecom rb against GPS, disciplining can be accomplished through
control of a biasing magnetic field.

Something like a cesium beam standard is able to internally cancel
most of these biases "under standard conditions".  A drift free
frequency source can also be constructed using rubidium, such as
rubidium fountains just as a secondary standard could be constructed
using cs-- like cs gas cell standards (such as the sa.45 CSAC).

[Hopefully I haven't mangled things].

On Mon, Nov 20, 2017 at 8:28 PM, Jerry Hancock <jerry@hanler.com> wrote: > Bob, I was referring to the rubidium standard of 6834682610.904 Hz. For some reason I thought it was closer to 9Ghz. > > I assume then rubidium standards oscillate (if that is the correct term) somewhere around that number but not exact or is it in the detection where things fall down? I think you are confused by the difference between primary and secondary standards. A typical rb gas cell is a secondary standard. Its exact frequency is distorted by a number of factors like gas pressure, interaction with the cell walls, and ambient magnetic fields which cannot be canceled by the design of the standard. This is why it is useful to discipline a telecom rb against GPS, disciplining can be accomplished through control of a biasing magnetic field. Something like a cesium beam standard is able to internally cancel most of these biases "under standard conditions". A drift free frequency source can also be constructed using rubidium, such as rubidium fountains just as a secondary standard could be constructed using cs-- like cs gas cell standards (such as the sa.45 CSAC). [Hopefully I haven't mangled things].
BK
Bob kb8tq
Mon, Nov 20, 2017 11:05 PM

Hi

It’s very much a “somewhere near that number” sort of thing with an Rb. The
“thing” you are looking at is quantum mechanical in nature. Unfortunately that
by its self does not make it perfect. A beam tube (as opposed to a gas cell)
isolates things better.

A 5061 is a beam tube device. A 5065 is gas cell based. It is very important to note that
accuracy and stability are two different things …. The beam tube is more accurate.
The gas cell is more stable (over some range of tau).

A normal Rb standard has a bit of this and that in the bulb. These other gasses
help in various ways. They each also add a bit of “pull” to the frequency one way
or the other. They get you away from your “magic number” but the benefits they
deliver are worth the trouble. The exact gas mix gets into the “secret sauce” of
the Rb manufacturer. They each optimize things a bit differently. The walls
of the bulb get into the act ….

Beam standards are actually a bit old these days. The more modern approach
would be a fountain (toss the ion straight up and let it fall back to you). An even
more modern approach would be a trapped ion standard. The amount of money
involved goes up dramatically with each of those steps. You get rid of this and
that subtle effect with each improvement. Accuracy gets better and better.

Lots of choices !!!

Bob

On Nov 20, 2017, at 3:28 PM, Jerry Hancock jerry@hanler.com wrote:

Bob, I was referring to the rubidium standard of 6834682610.904 Hz.  For some reason I thought it was closer to 9Ghz.

I assume then rubidium standards oscillate (if that is the correct term) somewhere around that number but not exact or is it in the detection where things fall down?

On Nov 20, 2017, at 11:40 AM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

There is no direct relation for an Rb to 10 MYz. Cs beam tubes are what have a direct relation.
Even then, the qualifier is “under standard conditions”. They are sensitive to magnetic field. Rb’s
also are sensitive to magnetic field. Both can be tuned by varying the field. In the case of an Rb
that also takes care of a multitude of other issues.

In the case of Rb, there is a distribution of cells coming out of the manufacturing process. Some
are pretty close to the “right” frequency. Others are way off (as in 100’s of KHz or more). All of them
are capable of meeting the required specs. DDS techniques allow those cells to be used in a
production part. That increases the yield and thus drops the production cost.

Since you now magically have a DDS in the Rb, you can do all sorts of interesting things. If you
suddenly need a 9.99900 MHz standard …. here it is … If you need to do temperature compensation
via a lookup table … it just takes a bit of testing and some code to make it happen. Indeed, the DDS
does also give you some issues. Without some sort of cleanup oscillator, you will have spurs and
phase noise on the output.

Lots of fun ….

Bob

On Nov 20, 2017, at 1:34 PM, Jerry Hancock jerry@hanler.com wrote:

I know this is going to sound dumb as I know many GPSDOs had rubidium oscillators in them.  I can see why, in that during holdover, they would tend to be more stable vs others, but given that there is a direct mathematical relationship between the rubidium frequency and potentially the 10Mhz desired output frequency, why do they have to be disciplined or better yet, what advantage does it bring?  Also, I can see how you discipline a DOCXO with the external voltage, how do you discipline a rubidium?  Pulse stretching?

I guess I don’t understand how the technology works, but it seems like an RF signal is swept that would be used to detect a dip at a pretty well defined frequency.  This dip can be used to discipline the oscillator to something like 9Ghz or a factor of what, 900+ times better than 10Mhz.  So wouldn’t that be able to get your desired 10Mhz to 10,000,000.001 or pretty much my level of measurement?  Or does is the dip not quite that precise?  If you can point me to a write-up on this I’ll go away.

Thanks to Gilbert for providing me with at least one rubidium oscillator that is working out of 5 though 2 others seems to stay locked for a few hours during my testing.

Jerry


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi It’s very much a “somewhere near that number” sort of thing with an Rb. The “thing” you are looking at is quantum mechanical in nature. Unfortunately that by its self does not make it perfect. A beam tube (as opposed to a gas cell) isolates things better. A 5061 is a beam tube device. A 5065 is gas cell based. It is very important to note that accuracy and stability are two different things …. The beam tube is more accurate. The gas cell is more stable (over some range of tau). A normal Rb standard has a bit of this and that in the bulb. These other gasses help in various ways. They each also add a bit of “pull” to the frequency one way or the other. They get you away from your “magic number” but the benefits they deliver are worth the trouble. The exact gas mix gets into the “secret sauce” of the Rb manufacturer. They each optimize things a bit differently. The walls of the bulb get into the act …. Beam standards are actually a bit old these days. The more modern approach would be a fountain (toss the ion straight up and let it fall back to you). An even more modern approach would be a trapped ion standard. The amount of money involved goes up dramatically with each of those steps. You get rid of this and that subtle effect with each improvement. Accuracy gets better and better. Lots of choices !!! Bob > On Nov 20, 2017, at 3:28 PM, Jerry Hancock <jerry@hanler.com> wrote: > > Bob, I was referring to the rubidium standard of 6834682610.904 Hz. For some reason I thought it was closer to 9Ghz. > > I assume then rubidium standards oscillate (if that is the correct term) somewhere around that number but not exact or is it in the detection where things fall down? > > > >> On Nov 20, 2017, at 11:40 AM, Bob kb8tq <kb8tq@n1k.org> wrote: >> >> Hi >> >> There is no direct relation for an Rb to 10 MYz. Cs beam tubes are what have a direct relation. >> Even then, the qualifier is “under standard conditions”. They are sensitive to magnetic field. Rb’s >> also are sensitive to magnetic field. Both can be tuned by varying the field. In the case of an Rb >> that also takes care of a multitude of other issues. >> >> In the case of Rb, there is a distribution of cells coming out of the manufacturing process. Some >> are pretty close to the “right” frequency. Others are way off (as in 100’s of KHz or more). All of them >> are capable of meeting the required specs. DDS techniques allow those cells to be used in a >> production part. That increases the yield and thus drops the production cost. >> >> Since you now magically have a DDS in the Rb, you can do all sorts of interesting things. If you >> suddenly need a 9.99900 MHz standard …. here it is … If you need to do temperature compensation >> via a lookup table … it just takes a bit of testing and some code to make it happen. Indeed, the DDS >> does also give you some issues. Without some sort of cleanup oscillator, you will have spurs and >> phase noise on the output. >> >> Lots of fun …. >> >> Bob >> >> >>> On Nov 20, 2017, at 1:34 PM, Jerry Hancock <jerry@hanler.com> wrote: >>> >>> I know this is going to sound dumb as I know many GPSDOs had rubidium oscillators in them. I can see why, in that during holdover, they would tend to be more stable vs others, but given that there is a direct mathematical relationship between the rubidium frequency and potentially the 10Mhz desired output frequency, why do they have to be disciplined or better yet, what advantage does it bring? Also, I can see how you discipline a DOCXO with the external voltage, how do you discipline a rubidium? Pulse stretching? >>> >>> I guess I don’t understand how the technology works, but it seems like an RF signal is swept that would be used to detect a dip at a pretty well defined frequency. This dip can be used to discipline the oscillator to something like 9Ghz or a factor of what, 900+ times better than 10Mhz. So wouldn’t that be able to get your desired 10Mhz to 10,000,000.001 or pretty much my level of measurement? Or does is the dip not quite that precise? If you can point me to a write-up on this I’ll go away. >>> >>> Thanks to Gilbert for providing me with at least one rubidium oscillator that is working out of 5 though 2 others seems to stay locked for a few hours during my testing. >>> >>> Jerry >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
JH
Jerry Hancock
Mon, Nov 20, 2017 11:31 PM

One step at a time.

2yrs ago when the time-bug hit, I had a crystal oscillator.  6 months later, DOCXO then GPSDO then Rubidium soon to be with GPSDO and there aren’t too many steps after that…

I also gave my brother the bug the other day…

On Nov 20, 2017, at 3:05 PM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

It’s very much a “somewhere near that number” sort of thing with an Rb. The
“thing” you are looking at is quantum mechanical in nature. Unfortunately that
by its self does not make it perfect. A beam tube (as opposed to a gas cell)
isolates things better.

A 5061 is a beam tube device. A 5065 is gas cell based. It is very important to note that
accuracy and stability are two different things …. The beam tube is more accurate.
The gas cell is more stable (over some range of tau).

A normal Rb standard has a bit of this and that in the bulb. These other gasses
help in various ways. They each also add a bit of “pull” to the frequency one way
or the other. They get you away from your “magic number” but the benefits they
deliver are worth the trouble. The exact gas mix gets into the “secret sauce” of
the Rb manufacturer. They each optimize things a bit differently. The walls
of the bulb get into the act ….

Beam standards are actually a bit old these days. The more modern approach
would be a fountain (toss the ion straight up and let it fall back to you). An even
more modern approach would be a trapped ion standard. The amount of money
involved goes up dramatically with each of those steps. You get rid of this and
that subtle effect with each improvement. Accuracy gets better and better.

Lots of choices !!!

Bob

On Nov 20, 2017, at 3:28 PM, Jerry Hancock jerry@hanler.com wrote:

Bob, I was referring to the rubidium standard of 6834682610.904 Hz.  For some reason I thought it was closer to 9Ghz.

I assume then rubidium standards oscillate (if that is the correct term) somewhere around that number but not exact or is it in the detection where things fall down?

On Nov 20, 2017, at 11:40 AM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

There is no direct relation for an Rb to 10 MYz. Cs beam tubes are what have a direct relation.
Even then, the qualifier is “under standard conditions”. They are sensitive to magnetic field. Rb’s
also are sensitive to magnetic field. Both can be tuned by varying the field. In the case of an Rb
that also takes care of a multitude of other issues.

In the case of Rb, there is a distribution of cells coming out of the manufacturing process. Some
are pretty close to the “right” frequency. Others are way off (as in 100’s of KHz or more). All of them
are capable of meeting the required specs. DDS techniques allow those cells to be used in a
production part. That increases the yield and thus drops the production cost.

Since you now magically have a DDS in the Rb, you can do all sorts of interesting things. If you
suddenly need a 9.99900 MHz standard …. here it is … If you need to do temperature compensation
via a lookup table … it just takes a bit of testing and some code to make it happen. Indeed, the DDS
does also give you some issues. Without some sort of cleanup oscillator, you will have spurs and
phase noise on the output.

Lots of fun ….

Bob

On Nov 20, 2017, at 1:34 PM, Jerry Hancock jerry@hanler.com wrote:

I know this is going to sound dumb as I know many GPSDOs had rubidium oscillators in them.  I can see why, in that during holdover, they would tend to be more stable vs others, but given that there is a direct mathematical relationship between the rubidium frequency and potentially the 10Mhz desired output frequency, why do they have to be disciplined or better yet, what advantage does it bring?  Also, I can see how you discipline a DOCXO with the external voltage, how do you discipline a rubidium?  Pulse stretching?

I guess I don’t understand how the technology works, but it seems like an RF signal is swept that would be used to detect a dip at a pretty well defined frequency.  This dip can be used to discipline the oscillator to something like 9Ghz or a factor of what, 900+ times better than 10Mhz.  So wouldn’t that be able to get your desired 10Mhz to 10,000,000.001 or pretty much my level of measurement?  Or does is the dip not quite that precise?  If you can point me to a write-up on this I’ll go away.

Thanks to Gilbert for providing me with at least one rubidium oscillator that is working out of 5 though 2 others seems to stay locked for a few hours during my testing.

Jerry


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

One step at a time. 2yrs ago when the time-bug hit, I had a crystal oscillator. 6 months later, DOCXO then GPSDO then Rubidium soon to be with GPSDO and there aren’t too many steps after that… I also gave my brother the bug the other day… > On Nov 20, 2017, at 3:05 PM, Bob kb8tq <kb8tq@n1k.org> wrote: > > Hi > > It’s very much a “somewhere near that number” sort of thing with an Rb. The > “thing” you are looking at is quantum mechanical in nature. Unfortunately that > by its self does not make it perfect. A beam tube (as opposed to a gas cell) > isolates things better. > > A 5061 is a beam tube device. A 5065 is gas cell based. It is very important to note that > accuracy and stability are two different things …. The beam tube is more accurate. > The gas cell is more stable (over some range of tau). > > A normal Rb standard has a bit of this and that in the bulb. These other gasses > help in various ways. They each also add a bit of “pull” to the frequency one way > or the other. They get you away from your “magic number” but the benefits they > deliver are worth the trouble. The exact gas mix gets into the “secret sauce” of > the Rb manufacturer. They each optimize things a bit differently. The walls > of the bulb get into the act …. > > Beam standards are actually a bit old these days. The more modern approach > would be a fountain (toss the ion straight up and let it fall back to you). An even > more modern approach would be a trapped ion standard. The amount of money > involved goes up dramatically with each of those steps. You get rid of this and > that subtle effect with each improvement. Accuracy gets better and better. > > Lots of choices !!! > > Bob > >> On Nov 20, 2017, at 3:28 PM, Jerry Hancock <jerry@hanler.com> wrote: >> >> Bob, I was referring to the rubidium standard of 6834682610.904 Hz. For some reason I thought it was closer to 9Ghz. >> >> I assume then rubidium standards oscillate (if that is the correct term) somewhere around that number but not exact or is it in the detection where things fall down? >> >> >> >>> On Nov 20, 2017, at 11:40 AM, Bob kb8tq <kb8tq@n1k.org> wrote: >>> >>> Hi >>> >>> There is no direct relation for an Rb to 10 MYz. Cs beam tubes are what have a direct relation. >>> Even then, the qualifier is “under standard conditions”. They are sensitive to magnetic field. Rb’s >>> also are sensitive to magnetic field. Both can be tuned by varying the field. In the case of an Rb >>> that also takes care of a multitude of other issues. >>> >>> In the case of Rb, there is a distribution of cells coming out of the manufacturing process. Some >>> are pretty close to the “right” frequency. Others are way off (as in 100’s of KHz or more). All of them >>> are capable of meeting the required specs. DDS techniques allow those cells to be used in a >>> production part. That increases the yield and thus drops the production cost. >>> >>> Since you now magically have a DDS in the Rb, you can do all sorts of interesting things. If you >>> suddenly need a 9.99900 MHz standard …. here it is … If you need to do temperature compensation >>> via a lookup table … it just takes a bit of testing and some code to make it happen. Indeed, the DDS >>> does also give you some issues. Without some sort of cleanup oscillator, you will have spurs and >>> phase noise on the output. >>> >>> Lots of fun …. >>> >>> Bob >>> >>> >>>> On Nov 20, 2017, at 1:34 PM, Jerry Hancock <jerry@hanler.com> wrote: >>>> >>>> I know this is going to sound dumb as I know many GPSDOs had rubidium oscillators in them. I can see why, in that during holdover, they would tend to be more stable vs others, but given that there is a direct mathematical relationship between the rubidium frequency and potentially the 10Mhz desired output frequency, why do they have to be disciplined or better yet, what advantage does it bring? Also, I can see how you discipline a DOCXO with the external voltage, how do you discipline a rubidium? Pulse stretching? >>>> >>>> I guess I don’t understand how the technology works, but it seems like an RF signal is swept that would be used to detect a dip at a pretty well defined frequency. This dip can be used to discipline the oscillator to something like 9Ghz or a factor of what, 900+ times better than 10Mhz. So wouldn’t that be able to get your desired 10Mhz to 10,000,000.001 or pretty much my level of measurement? Or does is the dip not quite that precise? If you can point me to a write-up on this I’ll go away. >>>> >>>> Thanks to Gilbert for providing me with at least one rubidium oscillator that is working out of 5 though 2 others seems to stay locked for a few hours during my testing. >>>> >>>> Jerry >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>> and follow the instructions there. >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BK
Bob kb8tq
Tue, Nov 21, 2017 12:09 AM

Hi

Ummm ….. errrr ….. multiple GPSDO’s …. L1/L2 GPSDO(s) …. Cs standard (s) … Maser(s) …. Ensembles of all of the above ….

There’s lots of steps still to take ….

Bob

On Nov 20, 2017, at 6:31 PM, Jerry Hancock jerry@hanler.com wrote:

One step at a time.

2yrs ago when the time-bug hit, I had a crystal oscillator.  6 months later, DOCXO then GPSDO then Rubidium soon to be with GPSDO and there aren’t too many steps after that…

I also gave my brother the bug the other day…

On Nov 20, 2017, at 3:05 PM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

It’s very much a “somewhere near that number” sort of thing with an Rb. The
“thing” you are looking at is quantum mechanical in nature. Unfortunately that
by its self does not make it perfect. A beam tube (as opposed to a gas cell)
isolates things better.

A 5061 is a beam tube device. A 5065 is gas cell based. It is very important to note that
accuracy and stability are two different things …. The beam tube is more accurate.
The gas cell is more stable (over some range of tau).

A normal Rb standard has a bit of this and that in the bulb. These other gasses
help in various ways. They each also add a bit of “pull” to the frequency one way
or the other. They get you away from your “magic number” but the benefits they
deliver are worth the trouble. The exact gas mix gets into the “secret sauce” of
the Rb manufacturer. They each optimize things a bit differently. The walls
of the bulb get into the act ….

Beam standards are actually a bit old these days. The more modern approach
would be a fountain (toss the ion straight up and let it fall back to you). An even
more modern approach would be a trapped ion standard. The amount of money
involved goes up dramatically with each of those steps. You get rid of this and
that subtle effect with each improvement. Accuracy gets better and better.

Lots of choices !!!

Bob

On Nov 20, 2017, at 3:28 PM, Jerry Hancock jerry@hanler.com wrote:

Bob, I was referring to the rubidium standard of 6834682610.904 Hz.  For some reason I thought it was closer to 9Ghz.

I assume then rubidium standards oscillate (if that is the correct term) somewhere around that number but not exact or is it in the detection where things fall down?

On Nov 20, 2017, at 11:40 AM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

There is no direct relation for an Rb to 10 MYz. Cs beam tubes are what have a direct relation.
Even then, the qualifier is “under standard conditions”. They are sensitive to magnetic field. Rb’s
also are sensitive to magnetic field. Both can be tuned by varying the field. In the case of an Rb
that also takes care of a multitude of other issues.

In the case of Rb, there is a distribution of cells coming out of the manufacturing process. Some
are pretty close to the “right” frequency. Others are way off (as in 100’s of KHz or more). All of them
are capable of meeting the required specs. DDS techniques allow those cells to be used in a
production part. That increases the yield and thus drops the production cost.

Since you now magically have a DDS in the Rb, you can do all sorts of interesting things. If you
suddenly need a 9.99900 MHz standard …. here it is … If you need to do temperature compensation
via a lookup table … it just takes a bit of testing and some code to make it happen. Indeed, the DDS
does also give you some issues. Without some sort of cleanup oscillator, you will have spurs and
phase noise on the output.

Lots of fun ….

Bob

On Nov 20, 2017, at 1:34 PM, Jerry Hancock jerry@hanler.com wrote:

I know this is going to sound dumb as I know many GPSDOs had rubidium oscillators in them.  I can see why, in that during holdover, they would tend to be more stable vs others, but given that there is a direct mathematical relationship between the rubidium frequency and potentially the 10Mhz desired output frequency, why do they have to be disciplined or better yet, what advantage does it bring?  Also, I can see how you discipline a DOCXO with the external voltage, how do you discipline a rubidium?  Pulse stretching?

I guess I don’t understand how the technology works, but it seems like an RF signal is swept that would be used to detect a dip at a pretty well defined frequency.  This dip can be used to discipline the oscillator to something like 9Ghz or a factor of what, 900+ times better than 10Mhz.  So wouldn’t that be able to get your desired 10Mhz to 10,000,000.001 or pretty much my level of measurement?  Or does is the dip not quite that precise?  If you can point me to a write-up on this I’ll go away.

Thanks to Gilbert for providing me with at least one rubidium oscillator that is working out of 5 though 2 others seems to stay locked for a few hours during my testing.

Jerry


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Ummm ….. errrr ….. multiple GPSDO’s …. L1/L2 GPSDO(s) …. Cs standard (s) … Maser(s) …. Ensembles of all of the above …. There’s *lots* of steps still to take …. Bob > On Nov 20, 2017, at 6:31 PM, Jerry Hancock <jerry@hanler.com> wrote: > > One step at a time. > > 2yrs ago when the time-bug hit, I had a crystal oscillator. 6 months later, DOCXO then GPSDO then Rubidium soon to be with GPSDO and there aren’t too many steps after that… > > I also gave my brother the bug the other day… > > > >> On Nov 20, 2017, at 3:05 PM, Bob kb8tq <kb8tq@n1k.org> wrote: >> >> Hi >> >> It’s very much a “somewhere near that number” sort of thing with an Rb. The >> “thing” you are looking at is quantum mechanical in nature. Unfortunately that >> by its self does not make it perfect. A beam tube (as opposed to a gas cell) >> isolates things better. >> >> A 5061 is a beam tube device. A 5065 is gas cell based. It is very important to note that >> accuracy and stability are two different things …. The beam tube is more accurate. >> The gas cell is more stable (over some range of tau). >> >> A normal Rb standard has a bit of this and that in the bulb. These other gasses >> help in various ways. They each also add a bit of “pull” to the frequency one way >> or the other. They get you away from your “magic number” but the benefits they >> deliver are worth the trouble. The exact gas mix gets into the “secret sauce” of >> the Rb manufacturer. They each optimize things a bit differently. The walls >> of the bulb get into the act …. >> >> Beam standards are actually a bit old these days. The more modern approach >> would be a fountain (toss the ion straight up and let it fall back to you). An even >> more modern approach would be a trapped ion standard. The amount of money >> involved goes up dramatically with each of those steps. You get rid of this and >> that subtle effect with each improvement. Accuracy gets better and better. >> >> Lots of choices !!! >> >> Bob >> >>> On Nov 20, 2017, at 3:28 PM, Jerry Hancock <jerry@hanler.com> wrote: >>> >>> Bob, I was referring to the rubidium standard of 6834682610.904 Hz. For some reason I thought it was closer to 9Ghz. >>> >>> I assume then rubidium standards oscillate (if that is the correct term) somewhere around that number but not exact or is it in the detection where things fall down? >>> >>> >>> >>>> On Nov 20, 2017, at 11:40 AM, Bob kb8tq <kb8tq@n1k.org> wrote: >>>> >>>> Hi >>>> >>>> There is no direct relation for an Rb to 10 MYz. Cs beam tubes are what have a direct relation. >>>> Even then, the qualifier is “under standard conditions”. They are sensitive to magnetic field. Rb’s >>>> also are sensitive to magnetic field. Both can be tuned by varying the field. In the case of an Rb >>>> that also takes care of a multitude of other issues. >>>> >>>> In the case of Rb, there is a distribution of cells coming out of the manufacturing process. Some >>>> are pretty close to the “right” frequency. Others are way off (as in 100’s of KHz or more). All of them >>>> are capable of meeting the required specs. DDS techniques allow those cells to be used in a >>>> production part. That increases the yield and thus drops the production cost. >>>> >>>> Since you now magically have a DDS in the Rb, you can do all sorts of interesting things. If you >>>> suddenly need a 9.99900 MHz standard …. here it is … If you need to do temperature compensation >>>> via a lookup table … it just takes a bit of testing and some code to make it happen. Indeed, the DDS >>>> does also give you some issues. Without some sort of cleanup oscillator, you will have spurs and >>>> phase noise on the output. >>>> >>>> Lots of fun …. >>>> >>>> Bob >>>> >>>> >>>>> On Nov 20, 2017, at 1:34 PM, Jerry Hancock <jerry@hanler.com> wrote: >>>>> >>>>> I know this is going to sound dumb as I know many GPSDOs had rubidium oscillators in them. I can see why, in that during holdover, they would tend to be more stable vs others, but given that there is a direct mathematical relationship between the rubidium frequency and potentially the 10Mhz desired output frequency, why do they have to be disciplined or better yet, what advantage does it bring? Also, I can see how you discipline a DOCXO with the external voltage, how do you discipline a rubidium? Pulse stretching? >>>>> >>>>> I guess I don’t understand how the technology works, but it seems like an RF signal is swept that would be used to detect a dip at a pretty well defined frequency. This dip can be used to discipline the oscillator to something like 9Ghz or a factor of what, 900+ times better than 10Mhz. So wouldn’t that be able to get your desired 10Mhz to 10,000,000.001 or pretty much my level of measurement? Or does is the dip not quite that precise? If you can point me to a write-up on this I’ll go away. >>>>> >>>>> Thanks to Gilbert for providing me with at least one rubidium oscillator that is working out of 5 though 2 others seems to stay locked for a few hours during my testing. >>>>> >>>>> Jerry >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@febo.com >>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>>> and follow the instructions there. >>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>> and follow the instructions there. >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
JH
Jerry Hancock
Tue, Nov 21, 2017 12:14 AM

Not to junk up the mailboxes, but I have the multiple GPSDOs.  Don’t know what you mean by L1/L2 GPSDO, is that a quality statement?

Also, what would the next step cost me for a Cesium Beam?  Roughly?  And what order of magnitude improvement would that be for the cost?  Have to do a cost/benefit analysis for the wife...

On Nov 20, 2017, at 4:09 PM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

Ummm ….. errrr ….. multiple GPSDO’s …. L1/L2 GPSDO(s) …. Cs standard (s) … Maser(s) …. Ensembles of all of the above ….

There’s lots of steps still to take ….

Bob

On Nov 20, 2017, at 6:31 PM, Jerry Hancock jerry@hanler.com wrote:

One step at a time.

2yrs ago when the time-bug hit, I had a crystal oscillator.  6 months later, DOCXO then GPSDO then Rubidium soon to be with GPSDO and there aren’t too many steps after that…

I also gave my brother the bug the other day…

On Nov 20, 2017, at 3:05 PM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

It’s very much a “somewhere near that number” sort of thing with an Rb. The
“thing” you are looking at is quantum mechanical in nature. Unfortunately that
by its self does not make it perfect. A beam tube (as opposed to a gas cell)
isolates things better.

A 5061 is a beam tube device. A 5065 is gas cell based. It is very important to note that
accuracy and stability are two different things …. The beam tube is more accurate.
The gas cell is more stable (over some range of tau).

A normal Rb standard has a bit of this and that in the bulb. These other gasses
help in various ways. They each also add a bit of “pull” to the frequency one way
or the other. They get you away from your “magic number” but the benefits they
deliver are worth the trouble. The exact gas mix gets into the “secret sauce” of
the Rb manufacturer. They each optimize things a bit differently. The walls
of the bulb get into the act ….

Beam standards are actually a bit old these days. The more modern approach
would be a fountain (toss the ion straight up and let it fall back to you). An even
more modern approach would be a trapped ion standard. The amount of money
involved goes up dramatically with each of those steps. You get rid of this and
that subtle effect with each improvement. Accuracy gets better and better.

Lots of choices !!!

Bob

On Nov 20, 2017, at 3:28 PM, Jerry Hancock jerry@hanler.com wrote:

Bob, I was referring to the rubidium standard of 6834682610.904 Hz.  For some reason I thought it was closer to 9Ghz.

I assume then rubidium standards oscillate (if that is the correct term) somewhere around that number but not exact or is it in the detection where things fall down?

On Nov 20, 2017, at 11:40 AM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

There is no direct relation for an Rb to 10 MYz. Cs beam tubes are what have a direct relation.
Even then, the qualifier is “under standard conditions”. They are sensitive to magnetic field. Rb’s
also are sensitive to magnetic field. Both can be tuned by varying the field. In the case of an Rb
that also takes care of a multitude of other issues.

In the case of Rb, there is a distribution of cells coming out of the manufacturing process. Some
are pretty close to the “right” frequency. Others are way off (as in 100’s of KHz or more). All of them
are capable of meeting the required specs. DDS techniques allow those cells to be used in a
production part. That increases the yield and thus drops the production cost.

Since you now magically have a DDS in the Rb, you can do all sorts of interesting things. If you
suddenly need a 9.99900 MHz standard …. here it is … If you need to do temperature compensation
via a lookup table … it just takes a bit of testing and some code to make it happen. Indeed, the DDS
does also give you some issues. Without some sort of cleanup oscillator, you will have spurs and
phase noise on the output.

Lots of fun ….

Bob

On Nov 20, 2017, at 1:34 PM, Jerry Hancock jerry@hanler.com wrote:

I know this is going to sound dumb as I know many GPSDOs had rubidium oscillators in them.  I can see why, in that during holdover, they would tend to be more stable vs others, but given that there is a direct mathematical relationship between the rubidium frequency and potentially the 10Mhz desired output frequency, why do they have to be disciplined or better yet, what advantage does it bring?  Also, I can see how you discipline a DOCXO with the external voltage, how do you discipline a rubidium?  Pulse stretching?

I guess I don’t understand how the technology works, but it seems like an RF signal is swept that would be used to detect a dip at a pretty well defined frequency.  This dip can be used to discipline the oscillator to something like 9Ghz or a factor of what, 900+ times better than 10Mhz.  So wouldn’t that be able to get your desired 10Mhz to 10,000,000.001 or pretty much my level of measurement?  Or does is the dip not quite that precise?  If you can point me to a write-up on this I’ll go away.

Thanks to Gilbert for providing me with at least one rubidium oscillator that is working out of 5 though 2 others seems to stay locked for a few hours during my testing.

Jerry


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Not to junk up the mailboxes, but I have the multiple GPSDOs. Don’t know what you mean by L1/L2 GPSDO, is that a quality statement? Also, what would the next step cost me for a Cesium Beam? Roughly? And what order of magnitude improvement would that be for the cost? Have to do a cost/benefit analysis for the wife... > On Nov 20, 2017, at 4:09 PM, Bob kb8tq <kb8tq@n1k.org> wrote: > > Hi > > Ummm ….. errrr ….. multiple GPSDO’s …. L1/L2 GPSDO(s) …. Cs standard (s) … Maser(s) …. Ensembles of all of the above …. > > There’s *lots* of steps still to take …. > > Bob > >> On Nov 20, 2017, at 6:31 PM, Jerry Hancock <jerry@hanler.com> wrote: >> >> One step at a time. >> >> 2yrs ago when the time-bug hit, I had a crystal oscillator. 6 months later, DOCXO then GPSDO then Rubidium soon to be with GPSDO and there aren’t too many steps after that… >> >> I also gave my brother the bug the other day… >> >> >> >>> On Nov 20, 2017, at 3:05 PM, Bob kb8tq <kb8tq@n1k.org> wrote: >>> >>> Hi >>> >>> It’s very much a “somewhere near that number” sort of thing with an Rb. The >>> “thing” you are looking at is quantum mechanical in nature. Unfortunately that >>> by its self does not make it perfect. A beam tube (as opposed to a gas cell) >>> isolates things better. >>> >>> A 5061 is a beam tube device. A 5065 is gas cell based. It is very important to note that >>> accuracy and stability are two different things …. The beam tube is more accurate. >>> The gas cell is more stable (over some range of tau). >>> >>> A normal Rb standard has a bit of this and that in the bulb. These other gasses >>> help in various ways. They each also add a bit of “pull” to the frequency one way >>> or the other. They get you away from your “magic number” but the benefits they >>> deliver are worth the trouble. The exact gas mix gets into the “secret sauce” of >>> the Rb manufacturer. They each optimize things a bit differently. The walls >>> of the bulb get into the act …. >>> >>> Beam standards are actually a bit old these days. The more modern approach >>> would be a fountain (toss the ion straight up and let it fall back to you). An even >>> more modern approach would be a trapped ion standard. The amount of money >>> involved goes up dramatically with each of those steps. You get rid of this and >>> that subtle effect with each improvement. Accuracy gets better and better. >>> >>> Lots of choices !!! >>> >>> Bob >>> >>>> On Nov 20, 2017, at 3:28 PM, Jerry Hancock <jerry@hanler.com> wrote: >>>> >>>> Bob, I was referring to the rubidium standard of 6834682610.904 Hz. For some reason I thought it was closer to 9Ghz. >>>> >>>> I assume then rubidium standards oscillate (if that is the correct term) somewhere around that number but not exact or is it in the detection where things fall down? >>>> >>>> >>>> >>>>> On Nov 20, 2017, at 11:40 AM, Bob kb8tq <kb8tq@n1k.org> wrote: >>>>> >>>>> Hi >>>>> >>>>> There is no direct relation for an Rb to 10 MYz. Cs beam tubes are what have a direct relation. >>>>> Even then, the qualifier is “under standard conditions”. They are sensitive to magnetic field. Rb’s >>>>> also are sensitive to magnetic field. Both can be tuned by varying the field. In the case of an Rb >>>>> that also takes care of a multitude of other issues. >>>>> >>>>> In the case of Rb, there is a distribution of cells coming out of the manufacturing process. Some >>>>> are pretty close to the “right” frequency. Others are way off (as in 100’s of KHz or more). All of them >>>>> are capable of meeting the required specs. DDS techniques allow those cells to be used in a >>>>> production part. That increases the yield and thus drops the production cost. >>>>> >>>>> Since you now magically have a DDS in the Rb, you can do all sorts of interesting things. If you >>>>> suddenly need a 9.99900 MHz standard …. here it is … If you need to do temperature compensation >>>>> via a lookup table … it just takes a bit of testing and some code to make it happen. Indeed, the DDS >>>>> does also give you some issues. Without some sort of cleanup oscillator, you will have spurs and >>>>> phase noise on the output. >>>>> >>>>> Lots of fun …. >>>>> >>>>> Bob >>>>> >>>>> >>>>>> On Nov 20, 2017, at 1:34 PM, Jerry Hancock <jerry@hanler.com> wrote: >>>>>> >>>>>> I know this is going to sound dumb as I know many GPSDOs had rubidium oscillators in them. I can see why, in that during holdover, they would tend to be more stable vs others, but given that there is a direct mathematical relationship between the rubidium frequency and potentially the 10Mhz desired output frequency, why do they have to be disciplined or better yet, what advantage does it bring? Also, I can see how you discipline a DOCXO with the external voltage, how do you discipline a rubidium? Pulse stretching? >>>>>> >>>>>> I guess I don’t understand how the technology works, but it seems like an RF signal is swept that would be used to detect a dip at a pretty well defined frequency. This dip can be used to discipline the oscillator to something like 9Ghz or a factor of what, 900+ times better than 10Mhz. So wouldn’t that be able to get your desired 10Mhz to 10,000,000.001 or pretty much my level of measurement? Or does is the dip not quite that precise? If you can point me to a write-up on this I’ll go away. >>>>>> >>>>>> Thanks to Gilbert for providing me with at least one rubidium oscillator that is working out of 5 though 2 others seems to stay locked for a few hours during my testing. >>>>>> >>>>>> Jerry >>>>>> _______________________________________________ >>>>>> time-nuts mailing list -- time-nuts@febo.com >>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>>>> and follow the instructions there. >>>>> >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@febo.com >>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>>> and follow the instructions there. >>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>> and follow the instructions there. >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BK
Bob kb8tq
Tue, Nov 21, 2017 1:18 AM

Hi

Most GPSDO’s run the GPS signal only on one “band” (L1). If you want to eliminate the errors in the ionosphere
correction process you go to.a double band (L1 and L2) GPS receiver. Since there are fewer potential errors in
the GPS signal, you may have fewer net errors in your GPSDO.

For the Cs - the issue is a stand alone reference. How can you be sure that GPS is not lying to you? The only way
to check something like that is with a stand alone reference. That’s somewhere between priceless and worthless
depending on which half of the family you happen to ask.

You can (or course) do a GPS disciplined Cs standard. That’s not easy to do, but some are attempting it ….

Lots to do…

Bob

On Nov 20, 2017, at 7:14 PM, Jerry Hancock jerry@hanler.com wrote:

Not to junk up the mailboxes, but I have the multiple GPSDOs.  Don’t know what you mean by L1/L2 GPSDO, is that a quality statement?

Also, what would the next step cost me for a Cesium Beam?  Roughly?  And what order of magnitude improvement would that be for the cost?  Have to do a cost/benefit analysis for the wife...

On Nov 20, 2017, at 4:09 PM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

Ummm ….. errrr ….. multiple GPSDO’s …. L1/L2 GPSDO(s) …. Cs standard (s) … Maser(s) …. Ensembles of all of the above ….

There’s lots of steps still to take ….

Bob

On Nov 20, 2017, at 6:31 PM, Jerry Hancock jerry@hanler.com wrote:

One step at a time.

2yrs ago when the time-bug hit, I had a crystal oscillator.  6 months later, DOCXO then GPSDO then Rubidium soon to be with GPSDO and there aren’t too many steps after that…

I also gave my brother the bug the other day…

On Nov 20, 2017, at 3:05 PM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

It’s very much a “somewhere near that number” sort of thing with an Rb. The
“thing” you are looking at is quantum mechanical in nature. Unfortunately that
by its self does not make it perfect. A beam tube (as opposed to a gas cell)
isolates things better.

A 5061 is a beam tube device. A 5065 is gas cell based. It is very important to note that
accuracy and stability are two different things …. The beam tube is more accurate.
The gas cell is more stable (over some range of tau).

A normal Rb standard has a bit of this and that in the bulb. These other gasses
help in various ways. They each also add a bit of “pull” to the frequency one way
or the other. They get you away from your “magic number” but the benefits they
deliver are worth the trouble. The exact gas mix gets into the “secret sauce” of
the Rb manufacturer. They each optimize things a bit differently. The walls
of the bulb get into the act ….

Beam standards are actually a bit old these days. The more modern approach
would be a fountain (toss the ion straight up and let it fall back to you). An even
more modern approach would be a trapped ion standard. The amount of money
involved goes up dramatically with each of those steps. You get rid of this and
that subtle effect with each improvement. Accuracy gets better and better.

Lots of choices !!!

Bob

On Nov 20, 2017, at 3:28 PM, Jerry Hancock jerry@hanler.com wrote:

Bob, I was referring to the rubidium standard of 6834682610.904 Hz.  For some reason I thought it was closer to 9Ghz.

I assume then rubidium standards oscillate (if that is the correct term) somewhere around that number but not exact or is it in the detection where things fall down?

On Nov 20, 2017, at 11:40 AM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

There is no direct relation for an Rb to 10 MYz. Cs beam tubes are what have a direct relation.
Even then, the qualifier is “under standard conditions”. They are sensitive to magnetic field. Rb’s
also are sensitive to magnetic field. Both can be tuned by varying the field. In the case of an Rb
that also takes care of a multitude of other issues.

In the case of Rb, there is a distribution of cells coming out of the manufacturing process. Some
are pretty close to the “right” frequency. Others are way off (as in 100’s of KHz or more). All of them
are capable of meeting the required specs. DDS techniques allow those cells to be used in a
production part. That increases the yield and thus drops the production cost.

Since you now magically have a DDS in the Rb, you can do all sorts of interesting things. If you
suddenly need a 9.99900 MHz standard …. here it is … If you need to do temperature compensation
via a lookup table … it just takes a bit of testing and some code to make it happen. Indeed, the DDS
does also give you some issues. Without some sort of cleanup oscillator, you will have spurs and
phase noise on the output.

Lots of fun ….

Bob

On Nov 20, 2017, at 1:34 PM, Jerry Hancock jerry@hanler.com wrote:

I know this is going to sound dumb as I know many GPSDOs had rubidium oscillators in them.  I can see why, in that during holdover, they would tend to be more stable vs others, but given that there is a direct mathematical relationship between the rubidium frequency and potentially the 10Mhz desired output frequency, why do they have to be disciplined or better yet, what advantage does it bring?  Also, I can see how you discipline a DOCXO with the external voltage, how do you discipline a rubidium?  Pulse stretching?

I guess I don’t understand how the technology works, but it seems like an RF signal is swept that would be used to detect a dip at a pretty well defined frequency.  This dip can be used to discipline the oscillator to something like 9Ghz or a factor of what, 900+ times better than 10Mhz.  So wouldn’t that be able to get your desired 10Mhz to 10,000,000.001 or pretty much my level of measurement?  Or does is the dip not quite that precise?  If you can point me to a write-up on this I’ll go away.

Thanks to Gilbert for providing me with at least one rubidium oscillator that is working out of 5 though 2 others seems to stay locked for a few hours during my testing.

Jerry


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Most GPSDO’s run the GPS signal only on one “band” (L1). If you want to eliminate the errors in the ionosphere correction process you go to.a double band (L1 and L2) GPS receiver. Since there are fewer potential errors in the GPS signal, you may have fewer net errors in your GPSDO. For the Cs - the issue is a stand alone reference. How can you be sure that GPS is not lying to you? The only way to check something like that is with a stand alone reference. That’s somewhere between priceless and worthless depending on which half of the family you happen to ask. You can (or course) do a GPS disciplined Cs standard. That’s not easy to do, but some are attempting it …. Lots to do… Bob > On Nov 20, 2017, at 7:14 PM, Jerry Hancock <jerry@hanler.com> wrote: > > Not to junk up the mailboxes, but I have the multiple GPSDOs. Don’t know what you mean by L1/L2 GPSDO, is that a quality statement? > > Also, what would the next step cost me for a Cesium Beam? Roughly? And what order of magnitude improvement would that be for the cost? Have to do a cost/benefit analysis for the wife... > > >> On Nov 20, 2017, at 4:09 PM, Bob kb8tq <kb8tq@n1k.org> wrote: >> >> Hi >> >> Ummm ….. errrr ….. multiple GPSDO’s …. L1/L2 GPSDO(s) …. Cs standard (s) … Maser(s) …. Ensembles of all of the above …. >> >> There’s *lots* of steps still to take …. >> >> Bob >> >>> On Nov 20, 2017, at 6:31 PM, Jerry Hancock <jerry@hanler.com> wrote: >>> >>> One step at a time. >>> >>> 2yrs ago when the time-bug hit, I had a crystal oscillator. 6 months later, DOCXO then GPSDO then Rubidium soon to be with GPSDO and there aren’t too many steps after that… >>> >>> I also gave my brother the bug the other day… >>> >>> >>> >>>> On Nov 20, 2017, at 3:05 PM, Bob kb8tq <kb8tq@n1k.org> wrote: >>>> >>>> Hi >>>> >>>> It’s very much a “somewhere near that number” sort of thing with an Rb. The >>>> “thing” you are looking at is quantum mechanical in nature. Unfortunately that >>>> by its self does not make it perfect. A beam tube (as opposed to a gas cell) >>>> isolates things better. >>>> >>>> A 5061 is a beam tube device. A 5065 is gas cell based. It is very important to note that >>>> accuracy and stability are two different things …. The beam tube is more accurate. >>>> The gas cell is more stable (over some range of tau). >>>> >>>> A normal Rb standard has a bit of this and that in the bulb. These other gasses >>>> help in various ways. They each also add a bit of “pull” to the frequency one way >>>> or the other. They get you away from your “magic number” but the benefits they >>>> deliver are worth the trouble. The exact gas mix gets into the “secret sauce” of >>>> the Rb manufacturer. They each optimize things a bit differently. The walls >>>> of the bulb get into the act …. >>>> >>>> Beam standards are actually a bit old these days. The more modern approach >>>> would be a fountain (toss the ion straight up and let it fall back to you). An even >>>> more modern approach would be a trapped ion standard. The amount of money >>>> involved goes up dramatically with each of those steps. You get rid of this and >>>> that subtle effect with each improvement. Accuracy gets better and better. >>>> >>>> Lots of choices !!! >>>> >>>> Bob >>>> >>>>> On Nov 20, 2017, at 3:28 PM, Jerry Hancock <jerry@hanler.com> wrote: >>>>> >>>>> Bob, I was referring to the rubidium standard of 6834682610.904 Hz. For some reason I thought it was closer to 9Ghz. >>>>> >>>>> I assume then rubidium standards oscillate (if that is the correct term) somewhere around that number but not exact or is it in the detection where things fall down? >>>>> >>>>> >>>>> >>>>>> On Nov 20, 2017, at 11:40 AM, Bob kb8tq <kb8tq@n1k.org> wrote: >>>>>> >>>>>> Hi >>>>>> >>>>>> There is no direct relation for an Rb to 10 MYz. Cs beam tubes are what have a direct relation. >>>>>> Even then, the qualifier is “under standard conditions”. They are sensitive to magnetic field. Rb’s >>>>>> also are sensitive to magnetic field. Both can be tuned by varying the field. In the case of an Rb >>>>>> that also takes care of a multitude of other issues. >>>>>> >>>>>> In the case of Rb, there is a distribution of cells coming out of the manufacturing process. Some >>>>>> are pretty close to the “right” frequency. Others are way off (as in 100’s of KHz or more). All of them >>>>>> are capable of meeting the required specs. DDS techniques allow those cells to be used in a >>>>>> production part. That increases the yield and thus drops the production cost. >>>>>> >>>>>> Since you now magically have a DDS in the Rb, you can do all sorts of interesting things. If you >>>>>> suddenly need a 9.99900 MHz standard …. here it is … If you need to do temperature compensation >>>>>> via a lookup table … it just takes a bit of testing and some code to make it happen. Indeed, the DDS >>>>>> does also give you some issues. Without some sort of cleanup oscillator, you will have spurs and >>>>>> phase noise on the output. >>>>>> >>>>>> Lots of fun …. >>>>>> >>>>>> Bob >>>>>> >>>>>> >>>>>>> On Nov 20, 2017, at 1:34 PM, Jerry Hancock <jerry@hanler.com> wrote: >>>>>>> >>>>>>> I know this is going to sound dumb as I know many GPSDOs had rubidium oscillators in them. I can see why, in that during holdover, they would tend to be more stable vs others, but given that there is a direct mathematical relationship between the rubidium frequency and potentially the 10Mhz desired output frequency, why do they have to be disciplined or better yet, what advantage does it bring? Also, I can see how you discipline a DOCXO with the external voltage, how do you discipline a rubidium? Pulse stretching? >>>>>>> >>>>>>> I guess I don’t understand how the technology works, but it seems like an RF signal is swept that would be used to detect a dip at a pretty well defined frequency. This dip can be used to discipline the oscillator to something like 9Ghz or a factor of what, 900+ times better than 10Mhz. So wouldn’t that be able to get your desired 10Mhz to 10,000,000.001 or pretty much my level of measurement? Or does is the dip not quite that precise? If you can point me to a write-up on this I’ll go away. >>>>>>> >>>>>>> Thanks to Gilbert for providing me with at least one rubidium oscillator that is working out of 5 though 2 others seems to stay locked for a few hours during my testing. >>>>>>> >>>>>>> Jerry >>>>>>> _______________________________________________ >>>>>>> time-nuts mailing list -- time-nuts@febo.com >>>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>>>>> and follow the instructions there. >>>>>> >>>>>> _______________________________________________ >>>>>> time-nuts mailing list -- time-nuts@febo.com >>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>>>> and follow the instructions there. >>>>> >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@febo.com >>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>>> and follow the instructions there. >>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>> and follow the instructions there. >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.