Luciano,
Not sure what you mean by "two measures made on the same DUT never
coincide".
The plotter program takes out wraps which since your phase gain is so
large are due to happen.
For lower Tau measurements just adjust the phase for close to equal and
run shorter plots.
These can be spliced into the longer Tau plot.
I routinely get repeatable plots on the same DUT with my DMTD setup.
Careful setup and understanding what you can and cannot accomplish is the
key.
Cheers,
Corby
Hi Corby,
I did several tests to verify the system noise floor using the same source for both input channels. I have seen that the result varies a lot as the difference in time between the two channels varies.
Moreover, if I do two successive tests without changing the parameters and the duts, the result can not be superimposed.
It may be that I make a mistake or do not take any parameter into account.
I also wonder if there is any commercial realization of the DMTD system.
Cheers,
Luciano
Da "time-nuts" time-nuts-bounces@febo.com
A time-nuts@febo.com
Cc
Data Thu, 18 Jan 2018 11:13:54 -0800
Oggetto [time-nuts] Time interval measurement vs dual mixer method
Luciano,
Not sure what you mean by "two measures made on the same DUT never
coincide".
The plotter program takes out wraps which since your phase gain is so
large are due to happen.
For lower Tau measurements just adjust the phase for close to equal and
run shorter plots.
These can be spliced into the longer Tau plot.
I routinely get repeatable plots on the same DUT with my DMTD setup.
Careful setup and understanding what you can and cannot accomplish is the
key.
Cheers,
Corby
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
Hi
The quality of the offset signal source matters in a DMTD system. If you are using
a “noisy” source you will have trouble with the data. You also will have trouble if the
offset signal is correlated in some way to one of the DUT signals.
Bob
On Jan 20, 2018, at 3:43 AM, timeok@timeok.it wrote:
Hi Corby,
I did several tests to verify the system noise floor using the same source for both input channels. I have seen that the result varies a lot as the difference in time between the two channels varies.
Moreover, if I do two successive tests without changing the parameters and the duts, the result can not be superimposed.
It may be that I make a mistake or do not take any parameter into account.
I also wonder if there is any commercial realization of the DMTD system.
Cheers,
Luciano
Da "time-nuts" time-nuts-bounces@febo.com
A time-nuts@febo.com
Cc
Data Thu, 18 Jan 2018 11:13:54 -0800
Oggetto [time-nuts] Time interval measurement vs dual mixer method
Luciano,
Not sure what you mean by "two measures made on the same DUT never
coincide".
The plotter program takes out wraps which since your phase gain is so
large are due to happen.
For lower Tau measurements just adjust the phase for close to equal and
run shorter plots.
These can be spliced into the longer Tau plot.
I routinely get repeatable plots on the same DUT with my DMTD setup.
Careful setup and understanding what you can and cannot accomplish is the
key.
Cheers,
Corby