Randy,
I doubt, that your hp unit is defective, as it might work under certain
circumstances, and also the FW version 8.2 vs. 9.2 will not make any
difference, as only a few pair of bytes are different. All versions of
the A5 board are designed to be completely interchangeable over all
versions of 3458A instruments.
I have another theory, which deals with the hardware version difference
between the agilent and the hp unit.
Your agilent unit might already have the 66547, full SMD board for A5,
whereas the hp will for sure have one of the old 66505/515, through-
hole boards, Rev A, B, or C. Illya has all of them pictured in his first
3458A repair blog.
All the old versions (A, B, C) have GPIB bus transceivers, TI 75ALS160
for the data bus, and NS 75161 for the handshake signals.
The new SMD board has TI 75ALS160 / TI 75ALS161, instead.
So, the handshake bus uses a different logic family for the new board,
and therefore, the logic levels / immunity might be slightly different.
My hp3458A has got the REV C A5 board, and it's from 2000.
Since 2009, I run a similar datalogging program for capturing LTZ1000
output versus time, each 4 seconds, on an old PC with WIN98, Turbo
Pascal, an XT bus GPIB card from CEC, with the NEC chip.
This program and hardware ran properly for years, (decades), until more
and more errors during the 24h acquisition runs occured, vaguely
reminding me of your trouble.
The output values were always correct, but frequently, 10sec time-outs
ocurred, giving zero readings in the datalog. This time-out is obviously
related to the handshaking mechanism.
Finally, I found out, that the old ATX PSU was failing, especially its
input snubber network was gone defect, which probably created big EMC
spikes.
After replacing the PSU, all these errors vanished completely.
My conclusion was, that the GPIB bus is susceptible to such EMC
disturbances, although it's solid TTL logic.
As obviously the handshake signals were affected only, their higher
susceptibility to disturbances may be caused by the difference in the
logic family of the transceiver IC. The agilent 3458A might be more
tolerant to that, having the ALS161 version, instead.
So you might also look for EMC problems in your lab, like SMPS, LED
lamps, or a defective PC power supply.
Maybe, these Prologix adaptors (clones?) have critical logic levels,
especially on the handshake bus, or an improper ground.
You might search for these signal problems directly on the GPIB cable.
Frank
Randy,
I put together a quick script for testing your connection. I have emailed
it to you but attached it here for others if needed. It will be easier than
running the other program.
Todd
On Sun, Dec 31, 2017 at 9:00 AM, Dr. Frank frank.stellmach@freenet.de
wrote:
Randy,
I doubt, that your hp unit is defective, as it might work under certain
circumstances, and also the FW version 8.2 vs. 9.2 will not make any
difference, as only a few pair of bytes are different. All versions of the
A5 board are designed to be completely interchangeable over all versions of
3458A instruments.
I have another theory, which deals with the hardware version difference
between the agilent and the hp unit.
Your agilent unit might already have the 66547, full SMD board for A5,
whereas the hp will for sure have one of the old 66505/515, through- hole
boards, Rev A, B, or C. Illya has all of them pictured in his first 3458A
repair blog.
All the old versions (A, B, C) have GPIB bus transceivers, TI 75ALS160
for the data bus, and NS 75161 for the handshake signals.
The new SMD board has TI 75ALS160 / TI 75ALS161, instead.
So, the handshake bus uses a different logic family for the new board, and
therefore, the logic levels / immunity might be slightly different.
My hp3458A has got the REV C A5 board, and it's from 2000.
Since 2009, I run a similar datalogging program for capturing LTZ1000
output versus time, each 4 seconds, on an old PC with WIN98, Turbo Pascal,
an XT bus GPIB card from CEC, with the NEC chip.
This program and hardware ran properly for years, (decades), until more
and more errors during the 24h acquisition runs occured, vaguely reminding
me of your trouble.
The output values were always correct, but frequently, 10sec time-outs
ocurred, giving zero readings in the datalog. This time-out is obviously
related to the handshaking mechanism.
Finally, I found out, that the old ATX PSU was failing, especially its
input snubber network was gone defect, which probably created big EMC
spikes.
After replacing the PSU, all these errors vanished completely.
My conclusion was, that the GPIB bus is susceptible to such EMC
disturbances, although it's solid TTL logic.
As obviously the handshake signals were affected only, their higher
susceptibility to disturbances may be caused by the difference in the logic
family of the transceiver IC. The agilent 3458A might be more tolerant to
that, having the ALS161 version, instead.
So you might also look for EMC problems in your lab, like SMPS, LED lamps,
or a defective PC power supply.
Maybe, these Prologix adaptors (clones?) have critical logic levels,
especially on the handshake bus, or an improper ground.
You might search for these signal problems directly on the GPIB cable.
Frank
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/volt-nuts
and follow the instructions there.
Todd,
The routine looks like this on the Output Console:
HP3458A
24.4
0
1
The 0 pops up after several seconds and the 1 after even more seconds
later. No other output after the 1.
The temperature reading is OK but no voltage readings.
Thanks for the assistance but I likely have a hardware issue. I did the
service note 13A test and the unit passed it fine. I will have to do more
troubleshooting on my own but I really appreciate the help.
Randy Evans.
On Sun, Dec 31, 2017 at 6:09 AM, Todd Micallef tmicallef@gmail.com wrote:
Randy,
I put together a quick script for testing your connection. I have emailed
it to you but attached it here for others if needed. It will be easier than
running the other program.
Todd
On Sun, Dec 31, 2017 at 9:00 AM, Dr. Frank frank.stellmach@freenet.de
wrote:
Randy,
I doubt, that your hp unit is defective, as it might work under certain
circumstances, and also the FW version 8.2 vs. 9.2 will not make any
difference, as only a few pair of bytes are different. All versions of
the
A5 board are designed to be completely interchangeable over all versions
of
3458A instruments.
I have another theory, which deals with the hardware version difference
between the agilent and the hp unit.
Your agilent unit might already have the 66547, full SMD board for A5,
whereas the hp will for sure have one of the old 66505/515, through- hole
boards, Rev A, B, or C. Illya has all of them pictured in his first 3458A
repair blog.
All the old versions (A, B, C) have GPIB bus transceivers, TI 75ALS160
for the data bus, and NS 75161 for the handshake signals.
The new SMD board has TI 75ALS160 / TI 75ALS161, instead.
So, the handshake bus uses a different logic family for the new board,
and
therefore, the logic levels / immunity might be slightly different.
My hp3458A has got the REV C A5 board, and it's from 2000.
Since 2009, I run a similar datalogging program for capturing LTZ1000
output versus time, each 4 seconds, on an old PC with WIN98, Turbo
Pascal,
an XT bus GPIB card from CEC, with the NEC chip.
This program and hardware ran properly for years, (decades), until more
and more errors during the 24h acquisition runs occured, vaguely
reminding
me of your trouble.
The output values were always correct, but frequently, 10sec time-outs
ocurred, giving zero readings in the datalog. This time-out is obviously
related to the handshaking mechanism.
Finally, I found out, that the old ATX PSU was failing, especially its
input snubber network was gone defect, which probably created big EMC
spikes.
After replacing the PSU, all these errors vanished completely.
My conclusion was, that the GPIB bus is susceptible to such EMC
disturbances, although it's solid TTL logic.
As obviously the handshake signals were affected only, their higher
susceptibility to disturbances may be caused by the difference in the
logic
family of the transceiver IC. The agilent 3458A might be more tolerant to
that, having the ALS161 version, instead.
So you might also look for EMC problems in your lab, like SMPS, LED
lamps,
or a defective PC power supply.
Maybe, these Prologix adaptors (clones?) have critical logic levels,
especially on the handshake bus, or an improper ground.
You might search for these signal problems directly on the GPIB cable.
Frank
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.