Trying to repair an old 5065A.
After a lot of problems with the Rubidium cavity restoration, it was time to startthe unit.
Lucklily, I did not have the Rubidium temperature controller PCB mounted,but the Power supply regulator malfunctioned and blew the AC mains fuse (0.5A).
Removing that board I discovered that the "24-32V" raw DC was some 40 Volts.
Yes - The unit was correctly configured for 230VAC. Even withsome load, the DC was still ~40V
Unable to find out why, I had to remove the original transformer and replace itwith a 22V unit after which the rectified DC was back to some 32 Volts.
(In Europe, there are seldom primarys with two windings 115/230 which madethe search extra tedious).
The 5065 was originally wired with internal battery backup as well asa mechanical clock. These options was removed since the batterieswere deceased and the clock driver flip-flop unobtanium.
After PSU board repair, all supply voltages are normal, the Rubidiumcavity warms up and the crystal oscillator and syntesizer seems to work.
Have not gotten further, just want to know what the h**k was going onwith the transformer...
Ulf - SM6GXV
Ulf
Perhaps you are reading peak VAC that is 1.414 X the RMS voltage typically
called out in manuals. I would email you directly but don't seem to have
your email. Everything points back to time-nuts.
Regards
Paul
WB8TSL
On Thu, Jul 27, 2017 at 1:24 PM, Ulf Kylenfall via time-nuts <
time-nuts@febo.com> wrote:
Trying to repair an old 5065A.
After a lot of problems with the Rubidium cavity restoration, it was time
to startthe unit.
Lucklily, I did not have the Rubidium temperature controller PCB
mounted,but the Power supply regulator malfunctioned and blew the AC mains
fuse (0.5A).
Removing that board I discovered that the "24-32V" raw DC was some 40
Volts.
Yes - The unit was correctly configured for 230VAC. Even withsome load,
the DC was still ~40V
Unable to find out why, I had to remove the original transformer and
replace itwith a 22V unit after which the rectified DC was back to some 32
Volts.
(In Europe, there are seldom primarys with two windings 115/230 which
madethe search extra tedious).
The 5065 was originally wired with internal battery backup as well asa
mechanical clock. These options was removed since the batterieswere
deceased and the clock driver flip-flop unobtanium.
After PSU board repair, all supply voltages are normal, the Rubidiumcavity
warms up and the crystal oscillator and syntesizer seems to work.
Have not gotten further, just want to know what the h**k was going onwith
the transformer...
Ulf - SM6GXV
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.
In message 675538475.428966.1501176257165@mail.yahoo.com, Ulf Kylenfall via t
ime-nuts writes:
Removing that board I discovered that the "24-32V" raw DC was some
40 Volts.
That's a bit on the high side, but not excessively so.
HP tended to let the linear regulators shave a big slice in
precision instruments, probably to make sure that absolutely
no ripple makes it through.
If I were you, I would ditch the transformer entirely, and go with
an external DC PSU and put a high-quality DC/DC converter in the
HP5065A
See:
http://phk.freebsd.dk/hacks/HP5065A/20150930_dcdc/index.html
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
Poul
I looked at your site and the power supply. That makes a lot of sense.
Though I have not lost my transformer as Ulf has at least I see an answer
that I can use. I would have come up with the same approach but you did the
thinking for me.
Thanks
Paul
WB8TSL
On Thu, Jul 27, 2017 at 3:58 PM, Poul-Henning Kamp phk@phk.freebsd.dk
wrote:
In message 675538475.428966.1501176257165@mail.yahoo.com, Ulf Kylenfall
via t
ime-nuts writes:
Removing that board I discovered that the "24-32V" raw DC was some
40 Volts.
That's a bit on the high side, but not excessively so.
HP tended to let the linear regulators shave a big slice in
precision instruments, probably to make sure that absolutely
no ripple makes it through.
If I were you, I would ditch the transformer entirely, and go with
an external DC PSU and put a high-quality DC/DC converter in the
HP5065A
See:
http://phk.freebsd.dk/hacks/HP5065A/20150930_dcdc/index.html
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.