volt-nuts@lists.febo.com

Discussion of precise voltage measurement

View all threads

Schematic error in Solartron 7150 Plus service manual

DC
David C. Partridge
Tue, Nov 9, 2010 10:58 AM

If you ever need to work on the Ohms circuit of a Solartron 7150 Plus, be warned, there is a schematic error in the service manual.  In diagram 2.1 (PCB1, Floating Analog), the top of resistor R22 should connect to the Source terminals of TR4, TR9 and the top of R26.  This "wire" should NOT connect to the gate of TR4 as it appears to do in the schematic.

I couldn't work out what was going on here for a while, and then the penny dropped!

The schematic in the original 7150 Service manual is correct.

Regards,
David Partridge

If you ever need to work on the Ohms circuit of a Solartron 7150 Plus, be warned, there is a schematic error in the service manual. In diagram 2.1 (PCB1, Floating Analog), the top of resistor R22 should connect to the Source terminals of TR4, TR9 and the top of R26. This "wire" should NOT connect to the gate of TR4 as it appears to do in the schematic. I couldn't work out what was going on here for a while, and then the penny dropped! The schematic in the original 7150 Service manual is correct. Regards, David Partridge
DC
David C. Partridge
Wed, Nov 10, 2010 5:34 PM

An updated version of the manual with schematic corrections has been provided by Nigel Clarke (GandalfG8@aol.com) and has been uploaded to Didier Juge's website <www.ko4bb.com>

Dave
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of David C. Partridge
Sent: 09 November 2010 10:59
To: volt-nuts@febo.com; TekScopes@yahoogroups.com; TekScopes2@yahoogroups.com; hp_agilent_equipment@yahoogroups.com
Subject: [volt-nuts] Schematic error in Solartron 7150 Plus service manual

If you ever need to work on the Ohms circuit of a Solartron 7150 Plus, be warned, there is a schematic error in the service manual.  In diagram 2.1 (PCB1, Floating Analog), the top of resistor R22 should connect to the Source terminals of TR4, TR9 and the top of R26.  This "wire" should NOT connect to the gate of TR4 as it appears to do in the schematic.

I couldn't work out what was going on here for a while, and then the penny dropped!

The schematic in the original 7150 Service manual is correct.

Regards,
David Partridge


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

An updated version of the manual with schematic corrections has been provided by Nigel Clarke (GandalfG8@aol.com) and has been uploaded to Didier Juge's website <www.ko4bb.com> Dave -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of David C. Partridge Sent: 09 November 2010 10:59 To: volt-nuts@febo.com; TekScopes@yahoogroups.com; TekScopes2@yahoogroups.com; hp_agilent_equipment@yahoogroups.com Subject: [volt-nuts] Schematic error in Solartron 7150 Plus service manual If you ever need to work on the Ohms circuit of a Solartron 7150 Plus, be warned, there is a schematic error in the service manual. In diagram 2.1 (PCB1, Floating Analog), the top of resistor R22 should connect to the Source terminals of TR4, TR9 and the top of R26. This "wire" should NOT connect to the gate of TR4 as it appears to do in the schematic. I couldn't work out what was going on here for a while, and then the penny dropped! The schematic in the original 7150 Service manual is correct. Regards, David Partridge _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.