Hello,

My name is fred, PA4TIM, I'm a volt-nut. Took a lot of therapy to
confess ;-)

I was already fascinated by calibrations because I restore measurement
equipment. I collected some old calibrators like an old one from Philips
using nixies as indicator, a Tek AC calibrator ment for internal factory
use.

But a few months ago a got a mail from a stranger who noticed my website
and asked if I was interested in the stuff from his companies old
calibrating lab (for free !!!) The lab was closed 25 years ago, but the
stuff stayed there unused. Now they needed the room and "found" this
stuff but almost nobody knew what it was so a lot was thrown away. He
thought that was a shame and asked if he was allowed to find a
collector. Al stuff including manuals and calibration history and in
state like new. But not used for 25 years and some stuff suffered from
sleeping or was maybe already death when parked away

It started with a Guildline 4 cell unit. In perefect condition. I had to
repair the heater and some minor aging things.

A tek 576 curve tracer came as a bonus with a box new Tek parts.

Then things speed up, if I wanted some more. A complete Fluke
calibration setup. To bad they tossed away all cables.
-332B voltage standard, worked like a charm after changing one cap.
-760A meter calibrator, worked for a week, now constant jumps in
protecting and the home fuse blowes even if I put the plug in powered of
so I think a Y capacitor is leak. I will find that, no problem (I hope)
-510A AC reference standard, dead batteries but I just removed them
because they are optional. I left in the option board. Still not sure if
that is wise.

  • 731A, also dead nicads. I removed them and put in a 12V voltage
    regulator. Also had to replace a paper-foil capacitor.
  • 750A voltage divider. There is room for two strange batteries but as
    far as I understand it also workes without as long as you are sure You
    do not over voltage
    -some HP stuff like a 20 penta Ohm elctro Ohmmeter/voltage source, a
    time standard, pindiode modulator and low distrotion sinewave generator.

But then two very important parts of the set up. A 720 kelvin Varley
divider. All see-through plastic isolator standoffs and centre coupling
parts of the switches are desintegrated. That will become a hell of a
job.

The biggest problem was my 845AB, they allready removed the nicads and
added two extra 10V zeners. Thick metal multi-amp ones.  The problem is
the manual is very limited. There are some scope pictures and a small
trouble seek section.
I had to replace allmost every cap, so did them all. Then I kept
problems with stability. Then I took my lab-supply and put 10V over the
capacitor. It drew about 30mA and was stable. Then I raised the voltage
and when the zener started conducting the 845 draw 600mA. I think the
10V mega-zeners started drawing so much current the transformer
saturated or at least power got unstable. Could not close the top lid,
as soon as it came near the transformer the meter went nuts. I now
placed a 7812 and it is stable over the whole range I think. At 1uV full
scale with shorted input and knob at zero the meter deviated plus/minus
0,1 uV max, most times about 0,05uV. I think that is not bad but this is
my only null-detector so I do not know for sure.

Now I have to find out how to make cables to connect everything, find a
way to determine what will be my  STANDARD VOLT. I have the calibrating
history upto 1988. I can measure the differences between the cells but
first have to find out the safest way, and if that is comform the old
papers I think I take that as a basis. Then adjust the calibrators to
that value. This will probably still be more accurate as the average
multi or bench meter. I have a 6,5 digit as best one.

Does some one has a better way ?

For my collection See www.pa4tim.nl Not all the fluke stuff is on it
yet.

73

Fred

Hello, My name is fred, PA4TIM, I'm a volt-nut. Took a lot of therapy to confess ;-) I was already fascinated by calibrations because I restore measurement equipment. I collected some old calibrators like an old one from Philips using nixies as indicator, a Tek AC calibrator ment for internal factory use. But a few months ago a got a mail from a stranger who noticed my website and asked if I was interested in the stuff from his companies old calibrating lab (for free !!!) The lab was closed 25 years ago, but the stuff stayed there unused. Now they needed the room and "found" this stuff but almost nobody knew what it was so a lot was thrown away. He thought that was a shame and asked if he was allowed to find a collector. Al stuff including manuals and calibration history and in state like new. But not used for 25 years and some stuff suffered from sleeping or was maybe already death when parked away It started with a Guildline 4 cell unit. In perefect condition. I had to repair the heater and some minor aging things. A tek 576 curve tracer came as a bonus with a box new Tek parts. Then things speed up, if I wanted some more. A complete Fluke calibration setup. To bad they tossed away all cables. -332B voltage standard, worked like a charm after changing one cap. -760A meter calibrator, worked for a week, now constant jumps in protecting and the home fuse blowes even if I put the plug in powered of so I think a Y capacitor is leak. I will find that, no problem (I hope) -510A AC reference standard, dead batteries but I just removed them because they are optional. I left in the option board. Still not sure if that is wise. - 731A, also dead nicads. I removed them and put in a 12V voltage regulator. Also had to replace a paper-foil capacitor. - 750A voltage divider. There is room for two strange batteries but as far as I understand it also workes without as long as you are sure You do not over voltage -some HP stuff like a 20 penta Ohm elctro Ohmmeter/voltage source, a time standard, pindiode modulator and low distrotion sinewave generator. But then two very important parts of the set up. A 720 kelvin Varley divider. All see-through plastic isolator standoffs and centre coupling parts of the switches are desintegrated. That will become a hell of a job. The biggest problem was my 845AB, they allready removed the nicads and added two extra 10V zeners. Thick metal multi-amp ones. The problem is the manual is very limited. There are some scope pictures and a small trouble seek section. I had to replace allmost every cap, so did them all. Then I kept problems with stability. Then I took my lab-supply and put 10V over the capacitor. It drew about 30mA and was stable. Then I raised the voltage and when the zener started conducting the 845 draw 600mA. I think the 10V mega-zeners started drawing so much current the transformer saturated or at least power got unstable. Could not close the top lid, as soon as it came near the transformer the meter went nuts. I now placed a 7812 and it is stable over the whole range I think. At 1uV full scale with shorted input and knob at zero the meter deviated plus/minus 0,1 uV max, most times about 0,05uV. I think that is not bad but this is my only null-detector so I do not know for sure. Now I have to find out how to make cables to connect everything, find a way to determine what will be my STANDARD VOLT. I have the calibrating history upto 1988. I can measure the differences between the cells but first have to find out the safest way, and if that is comform the old papers I think I take that as a basis. Then adjust the calibrators to that value. This will probably still be more accurate as the average multi or bench meter. I have a 6,5 digit as best one. Does some one has a better way ? For my collection See www.pa4tim.nl Not all the fluke stuff is on it yet. 73 Fred
MG
Marv Gozum @ JHN
Tue, Apr 26, 2011 7:10 PM

Such luck!  Welcome, things couldn't be better for you, for this type
of fascination.

At 02:51 PM 4/26/2011, Fred wrote:

But a few months ago a got a mail from a stranger who noticed my website
and asked if I was interested in the stuff from his companies old
calibrating lab (for free !!!) The lab was closed 25 years ago, but the
stuff stayed there unused. Now they needed the room and "found" this
stuff but almost nobody knew what it was so a lot was thrown away. He
thought that was a shame and asked if he was allowed to find a
collector. Al stuff including manuals and calibration history and in
state like new. But not used for 25 years and some stuff suffered from
sleeping or was maybe already death when parked away

For my collection See www.pa4tim.nl Not all the fluke stuff is on it
yet.

73

Fred

Best Wishes,

Marv Gozum
Philadelphia

Such luck! Welcome, things couldn't be better for you, for this type of fascination. At 02:51 PM 4/26/2011, Fred wrote: >But a few months ago a got a mail from a stranger who noticed my website >and asked if I was interested in the stuff from his companies old >calibrating lab (for free !!!) The lab was closed 25 years ago, but the >stuff stayed there unused. Now they needed the room and "found" this >stuff but almost nobody knew what it was so a lot was thrown away. He >thought that was a shame and asked if he was allowed to find a >collector. Al stuff including manuals and calibration history and in >state like new. But not used for 25 years and some stuff suffered from >sleeping or was maybe already death when parked away > >For my collection See www.pa4tim.nl Not all the fluke stuff is on it >yet. > >73 > >Fred Best Wishes, Marv Gozum Philadelphia
W
WB6BNQ
Tue, Apr 26, 2011 8:20 PM

Hi Fred,

Looks like you got lucky with such a nice collection being offered.

I am confused on why additional zeners would be added to the 731 and 845.  Both
were made to work without batteries.

The 731, if it is in good shape, is an extremely stable reference source.  Well
able to transfer a known value from a competent calibration Lab to use for
establishing the value of your Guildline cells.

As I am sure you know, the 845 is the only high impedance DC NULL meter available
that was worth a damn.  The first mains power transformer (and battery with its
charge circuit) is just to supply a power source for the 84 hertz transistor
inverter to the neon's.  That same inverter is also used to power the all of the
circuitry.  That is part of what provided the extremely high input isolation.
The noise you see on the 1 uv scale is normal.

The 720 switch problem sounds interesting, I wonder what caused the switches to
be so bad ?  In a clean environment there should be no reason for them to have
disintegrated.

The Guildline was a very nice catch.  Have you had an opportunity to determine if
the cells are in good shape ?  Hopefully nobody loaded them with a low
resistance, that would have messed them up for sure.  In conjunction with the
845, 750 and the 731 you have yourself a nice voltage reference system.

As for cables, well, I would not worry about that.  Just use normal copper wire
and you will not go wrong.

From your comments I am wondering if you have the full manuals for the 731 and

845 ?

73....Bill....WB6BNQ

Fred wrote:

Hello,

My name is fred, PA4TIM, I'm a volt-nut. Took a lot of therapy to
confess ;-)

I was already fascinated by calibrations because I restore measurement
equipment. I collected some old calibrators like an old one from Philips
using nixies as indicator, a Tek AC calibrator ment for internal factory
use.

But a few months ago a got a mail from a stranger who noticed my website
and asked if I was interested in the stuff from his companies old
calibrating lab (for free !!!) The lab was closed 25 years ago, but the
stuff stayed there unused. Now they needed the room and "found" this
stuff but almost nobody knew what it was so a lot was thrown away. He
thought that was a shame and asked if he was allowed to find a
collector. Al stuff including manuals and calibration history and in
state like new. But not used for 25 years and some stuff suffered from
sleeping or was maybe already death when parked away

It started with a Guildline 4 cell unit. In perefect condition. I had to
repair the heater and some minor aging things.

A tek 576 curve tracer came as a bonus with a box new Tek parts.

Then things speed up, if I wanted some more. A complete Fluke
calibration setup. To bad they tossed away all cables.
-332B voltage standard, worked like a charm after changing one cap.
-760A meter calibrator, worked for a week, now constant jumps in
protecting and the home fuse blowes even if I put the plug in powered of
so I think a Y capacitor is leak. I will find that, no problem (I hope)
-510A AC reference standard, dead batteries but I just removed them
because they are optional. I left in the option board. Still not sure if
that is wise.

  • 731A, also dead nicads. I removed them and put in a 12V voltage
    regulator. Also had to replace a paper-foil capacitor.
  • 750A voltage divider. There is room for two strange batteries but as
    far as I understand it also workes without as long as you are sure You
    do not over voltage
    -some HP stuff like a 20 penta Ohm elctro Ohmmeter/voltage source, a
    time standard, pindiode modulator and low distrotion sinewave generator.

But then two very important parts of the set up. A 720 kelvin Varley
divider. All see-through plastic isolator standoffs and centre coupling
parts of the switches are desintegrated. That will become a hell of a
job.

The biggest problem was my 845AB, they allready removed the nicads and
added two extra 10V zeners. Thick metal multi-amp ones.  The problem is
the manual is very limited. There are some scope pictures and a small
trouble seek section.
I had to replace allmost every cap, so did them all. Then I kept
problems with stability. Then I took my lab-supply and put 10V over the
capacitor. It drew about 30mA and was stable. Then I raised the voltage
and when the zener started conducting the 845 draw 600mA. I think the
10V mega-zeners started drawing so much current the transformer
saturated or at least power got unstable. Could not close the top lid,
as soon as it came near the transformer the meter went nuts. I now
placed a 7812 and it is stable over the whole range I think. At 1uV full
scale with shorted input and knob at zero the meter deviated plus/minus
0,1 uV max, most times about 0,05uV. I think that is not bad but this is
my only null-detector so I do not know for sure.

Now I have to find out how to make cables to connect everything, find a
way to determine what will be my  STANDARD VOLT. I have the calibrating
history upto 1988. I can measure the differences between the cells but
first have to find out the safest way, and if that is comform the old
papers I think I take that as a basis. Then adjust the calibrators to
that value. This will probably still be more accurate as the average
multi or bench meter. I have a 6,5 digit as best one.

Does some one has a better way ?

For my collection See www.pa4tim.nl Not all the fluke stuff is on it
yet.

73

Fred


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi Fred, Looks like you got lucky with such a nice collection being offered. I am confused on why additional zeners would be added to the 731 and 845. Both were made to work without batteries. The 731, if it is in good shape, is an extremely stable reference source. Well able to transfer a known value from a competent calibration Lab to use for establishing the value of your Guildline cells. As I am sure you know, the 845 is the only high impedance DC NULL meter available that was worth a damn. The first mains power transformer (and battery with its charge circuit) is just to supply a power source for the 84 hertz transistor inverter to the neon's. That same inverter is also used to power the all of the circuitry. That is part of what provided the extremely high input isolation. The noise you see on the 1 uv scale is normal. The 720 switch problem sounds interesting, I wonder what caused the switches to be so bad ? In a clean environment there should be no reason for them to have disintegrated. The Guildline was a very nice catch. Have you had an opportunity to determine if the cells are in good shape ? Hopefully nobody loaded them with a low resistance, that would have messed them up for sure. In conjunction with the 845, 750 and the 731 you have yourself a nice voltage reference system. As for cables, well, I would not worry about that. Just use normal copper wire and you will not go wrong. >From your comments I am wondering if you have the full manuals for the 731 and 845 ? 73....Bill....WB6BNQ Fred wrote: > Hello, > > My name is fred, PA4TIM, I'm a volt-nut. Took a lot of therapy to > confess ;-) > > I was already fascinated by calibrations because I restore measurement > equipment. I collected some old calibrators like an old one from Philips > using nixies as indicator, a Tek AC calibrator ment for internal factory > use. > > But a few months ago a got a mail from a stranger who noticed my website > and asked if I was interested in the stuff from his companies old > calibrating lab (for free !!!) The lab was closed 25 years ago, but the > stuff stayed there unused. Now they needed the room and "found" this > stuff but almost nobody knew what it was so a lot was thrown away. He > thought that was a shame and asked if he was allowed to find a > collector. Al stuff including manuals and calibration history and in > state like new. But not used for 25 years and some stuff suffered from > sleeping or was maybe already death when parked away > > It started with a Guildline 4 cell unit. In perefect condition. I had to > repair the heater and some minor aging things. > > A tek 576 curve tracer came as a bonus with a box new Tek parts. > > Then things speed up, if I wanted some more. A complete Fluke > calibration setup. To bad they tossed away all cables. > -332B voltage standard, worked like a charm after changing one cap. > -760A meter calibrator, worked for a week, now constant jumps in > protecting and the home fuse blowes even if I put the plug in powered of > so I think a Y capacitor is leak. I will find that, no problem (I hope) > -510A AC reference standard, dead batteries but I just removed them > because they are optional. I left in the option board. Still not sure if > that is wise. > - 731A, also dead nicads. I removed them and put in a 12V voltage > regulator. Also had to replace a paper-foil capacitor. > - 750A voltage divider. There is room for two strange batteries but as > far as I understand it also workes without as long as you are sure You > do not over voltage > -some HP stuff like a 20 penta Ohm elctro Ohmmeter/voltage source, a > time standard, pindiode modulator and low distrotion sinewave generator. > > But then two very important parts of the set up. A 720 kelvin Varley > divider. All see-through plastic isolator standoffs and centre coupling > parts of the switches are desintegrated. That will become a hell of a > job. > > The biggest problem was my 845AB, they allready removed the nicads and > added two extra 10V zeners. Thick metal multi-amp ones. The problem is > the manual is very limited. There are some scope pictures and a small > trouble seek section. > I had to replace allmost every cap, so did them all. Then I kept > problems with stability. Then I took my lab-supply and put 10V over the > capacitor. It drew about 30mA and was stable. Then I raised the voltage > and when the zener started conducting the 845 draw 600mA. I think the > 10V mega-zeners started drawing so much current the transformer > saturated or at least power got unstable. Could not close the top lid, > as soon as it came near the transformer the meter went nuts. I now > placed a 7812 and it is stable over the whole range I think. At 1uV full > scale with shorted input and knob at zero the meter deviated plus/minus > 0,1 uV max, most times about 0,05uV. I think that is not bad but this is > my only null-detector so I do not know for sure. > > Now I have to find out how to make cables to connect everything, find a > way to determine what will be my STANDARD VOLT. I have the calibrating > history upto 1988. I can measure the differences between the cells but > first have to find out the safest way, and if that is comform the old > papers I think I take that as a basis. Then adjust the calibrators to > that value. This will probably still be more accurate as the average > multi or bench meter. I have a 6,5 digit as best one. > > Does some one has a better way ? > > > For my collection See www.pa4tim.nl Not all the fluke stuff is on it > yet. > > 73 > > Fred > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
F
Fred
Tue, Apr 26, 2011 10:18 PM

I think I have the manuals complete, they are original and I do not miss
pages but maybe there is also a real extended service manual for the
845AB.
In my schematic there is a A and a AB powersupply schematic. The A has a
zener and no battery. They removed the battery in mine, that was also
written on the cabinet and placed two zeners. I think they did not get
it right anymore and then bought the Fluke 8500 that was part of the
setup (in the drawings I had in the documentation it was coupled to the
750)

Tonight I did some first measurement. The Cells are now about a month or
two operational and heated. They are very carefull moved to a special
made shelf two weeks ago. But now I have a problem. It's like the man
with just 1 clock, he think he does know the time, the man with two
clocks has a problem. I have 4 DC calibrators and 4 Weston cells so I'm
lost :-)

I first measured 1 cell from the Guildline using the 845 and the 731. I
turned the 731 dial until I had a zero in the 3uV range and allmost zero
in the 1uV range. I took that value as a reference. Following the trend
in the graphs from this cell (from 1977 upto 1987) the value could be
realistic. The oven temp is 30.07 degrees. The philips was about plus
250uV, the 332 was minus 160uV. But if I connect them through the 750 I
get a 25uV difference. I have not tried to calibrate it, it is always
kept in the same clean room and never messed around so it will be
probably better if I leave it like this because the 720 is not usefull
and needed for calibration.

The 720 is strange, it is clean as a baby inside-out. It is not been
dropped or so. But all the plastic is cracked and broken. If you touch
it it crumbles. Maybe someone back then cleaned it with some solvent.
It's a pitty because it is a beautifull instrument. I must find some
teflon or poly-ethilene to make new ones (I have a lath and a mill but
is is a lot of work) Some are possible repairable using 2K-epoxy. One
switch, number 1 from right, is very bad, I had to take it out because
the centre-follower is crumbled. The others still work and are in place
but the studs have cracks and pieces allready missing.

73

Fred

WB6BNQ schreef op di 26-04-2011 om 13:20 [-0700]:

Hi Fred,

Looks like you got lucky with such a nice collection being offered.

I am confused on why additional zeners would be added to the 731 and 845.  Both
were made to work without batteries.

The 731, if it is in good shape, is an extremely stable reference source.  Well
able to transfer a known value from a competent calibration Lab to use for
establishing the value of your Guildline cells.

As I am sure you know, the 845 is the only high impedance DC NULL meter available
that was worth a damn.  The first mains power transformer (and battery with its
charge circuit) is just to supply a power source for the 84 hertz transistor
inverter to the neon's.  That same inverter is also used to power the all of the
circuitry.  That is part of what provided the extremely high input isolation.
The noise you see on the 1 uv scale is normal.

The 720 switch problem sounds interesting, I wonder what caused the switches to
be so bad ?  In a clean environment there should be no reason for them to have
disintegrated.

The Guildline was a very nice catch.  Have you had an opportunity to determine if
the cells are in good shape ?  Hopefully nobody loaded them with a low
resistance, that would have messed them up for sure.  In conjunction with the
845, 750 and the 731 you have yourself a nice voltage reference system.

As for cables, well, I would not worry about that.  Just use normal copper wire
and you will not go wrong.

From your comments I am wondering if you have the full manuals for the 731 and
845 ?

73....Bill....WB6BNQ

Fred wrote:

Hello,

My name is fred, PA4TIM, I'm a volt-nut. Took a lot of therapy to
confess ;-)

I was already fascinated by calibrations because I restore measurement
equipment. I collected some old calibrators like an old one from Philips
using nixies as indicator, a Tek AC calibrator ment for internal factory
use.

But a few months ago a got a mail from a stranger who noticed my website
and asked if I was interested in the stuff from his companies old
calibrating lab (for free !!!) The lab was closed 25 years ago, but the
stuff stayed there unused. Now they needed the room and "found" this
stuff but almost nobody knew what it was so a lot was thrown away. He
thought that was a shame and asked if he was allowed to find a
collector. Al stuff including manuals and calibration history and in
state like new. But not used for 25 years and some stuff suffered from
sleeping or was maybe already death when parked away

It started with a Guildline 4 cell unit. In perefect condition. I had to
repair the heater and some minor aging things.

A tek 576 curve tracer came as a bonus with a box new Tek parts.

Then things speed up, if I wanted some more. A complete Fluke
calibration setup. To bad they tossed away all cables.
-332B voltage standard, worked like a charm after changing one cap.
-760A meter calibrator, worked for a week, now constant jumps in
protecting and the home fuse blowes even if I put the plug in powered of
so I think a Y capacitor is leak. I will find that, no problem (I hope)
-510A AC reference standard, dead batteries but I just removed them
because they are optional. I left in the option board. Still not sure if
that is wise.

  • 731A, also dead nicads. I removed them and put in a 12V voltage
    regulator. Also had to replace a paper-foil capacitor.
  • 750A voltage divider. There is room for two strange batteries but as
    far as I understand it also workes without as long as you are sure You
    do not over voltage
    -some HP stuff like a 20 penta Ohm elctro Ohmmeter/voltage source, a
    time standard, pindiode modulator and low distrotion sinewave generator.

But then two very important parts of the set up. A 720 kelvin Varley
divider. All see-through plastic isolator standoffs and centre coupling
parts of the switches are desintegrated. That will become a hell of a
job.

The biggest problem was my 845AB, they allready removed the nicads and
added two extra 10V zeners. Thick metal multi-amp ones.  The problem is
the manual is very limited. There are some scope pictures and a small
trouble seek section.
I had to replace allmost every cap, so did them all. Then I kept
problems with stability. Then I took my lab-supply and put 10V over the
capacitor. It drew about 30mA and was stable. Then I raised the voltage
and when the zener started conducting the 845 draw 600mA. I think the
10V mega-zeners started drawing so much current the transformer
saturated or at least power got unstable. Could not close the top lid,
as soon as it came near the transformer the meter went nuts. I now
placed a 7812 and it is stable over the whole range I think. At 1uV full
scale with shorted input and knob at zero the meter deviated plus/minus
0,1 uV max, most times about 0,05uV. I think that is not bad but this is
my only null-detector so I do not know for sure.

Now I have to find out how to make cables to connect everything, find a
way to determine what will be my  STANDARD VOLT. I have the calibrating
history upto 1988. I can measure the differences between the cells but
first have to find out the safest way, and if that is comform the old
papers I think I take that as a basis. Then adjust the calibrators to
that value. This will probably still be more accurate as the average
multi or bench meter. I have a 6,5 digit as best one.

Does some one has a better way ?

For my collection See www.pa4tim.nl Not all the fluke stuff is on it
yet.

73

Fred


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

I think I have the manuals complete, they are original and I do not miss pages but maybe there is also a real extended service manual for the 845AB. In my schematic there is a A and a AB powersupply schematic. The A has a zener and no battery. They removed the battery in mine, that was also written on the cabinet and placed two zeners. I think they did not get it right anymore and then bought the Fluke 8500 that was part of the setup (in the drawings I had in the documentation it was coupled to the 750) Tonight I did some first measurement. The Cells are now about a month or two operational and heated. They are very carefull moved to a special made shelf two weeks ago. But now I have a problem. It's like the man with just 1 clock, he think he does know the time, the man with two clocks has a problem. I have 4 DC calibrators and 4 Weston cells so I'm lost :-) I first measured 1 cell from the Guildline using the 845 and the 731. I turned the 731 dial until I had a zero in the 3uV range and allmost zero in the 1uV range. I took that value as a reference. Following the trend in the graphs from this cell (from 1977 upto 1987) the value could be realistic. The oven temp is 30.07 degrees. The philips was about plus 250uV, the 332 was minus 160uV. But if I connect them through the 750 I get a 25uV difference. I have not tried to calibrate it, it is always kept in the same clean room and never messed around so it will be probably better if I leave it like this because the 720 is not usefull and needed for calibration. The 720 is strange, it is clean as a baby inside-out. It is not been dropped or so. But all the plastic is cracked and broken. If you touch it it crumbles. Maybe someone back then cleaned it with some solvent. It's a pitty because it is a beautifull instrument. I must find some teflon or poly-ethilene to make new ones (I have a lath and a mill but is is a lot of work) Some are possible repairable using 2K-epoxy. One switch, number 1 from right, is very bad, I had to take it out because the centre-follower is crumbled. The others still work and are in place but the studs have cracks and pieces allready missing. 73 Fred WB6BNQ schreef op di 26-04-2011 om 13:20 [-0700]: > Hi Fred, > > Looks like you got lucky with such a nice collection being offered. > > I am confused on why additional zeners would be added to the 731 and 845. Both > were made to work without batteries. > > The 731, if it is in good shape, is an extremely stable reference source. Well > able to transfer a known value from a competent calibration Lab to use for > establishing the value of your Guildline cells. > > As I am sure you know, the 845 is the only high impedance DC NULL meter available > that was worth a damn. The first mains power transformer (and battery with its > charge circuit) is just to supply a power source for the 84 hertz transistor > inverter to the neon's. That same inverter is also used to power the all of the > circuitry. That is part of what provided the extremely high input isolation. > The noise you see on the 1 uv scale is normal. > > The 720 switch problem sounds interesting, I wonder what caused the switches to > be so bad ? In a clean environment there should be no reason for them to have > disintegrated. > > The Guildline was a very nice catch. Have you had an opportunity to determine if > the cells are in good shape ? Hopefully nobody loaded them with a low > resistance, that would have messed them up for sure. In conjunction with the > 845, 750 and the 731 you have yourself a nice voltage reference system. > > As for cables, well, I would not worry about that. Just use normal copper wire > and you will not go wrong. > > From your comments I am wondering if you have the full manuals for the 731 and > 845 ? > > 73....Bill....WB6BNQ > > > Fred wrote: > > > Hello, > > > > My name is fred, PA4TIM, I'm a volt-nut. Took a lot of therapy to > > confess ;-) > > > > I was already fascinated by calibrations because I restore measurement > > equipment. I collected some old calibrators like an old one from Philips > > using nixies as indicator, a Tek AC calibrator ment for internal factory > > use. > > > > But a few months ago a got a mail from a stranger who noticed my website > > and asked if I was interested in the stuff from his companies old > > calibrating lab (for free !!!) The lab was closed 25 years ago, but the > > stuff stayed there unused. Now they needed the room and "found" this > > stuff but almost nobody knew what it was so a lot was thrown away. He > > thought that was a shame and asked if he was allowed to find a > > collector. Al stuff including manuals and calibration history and in > > state like new. But not used for 25 years and some stuff suffered from > > sleeping or was maybe already death when parked away > > > > It started with a Guildline 4 cell unit. In perefect condition. I had to > > repair the heater and some minor aging things. > > > > A tek 576 curve tracer came as a bonus with a box new Tek parts. > > > > Then things speed up, if I wanted some more. A complete Fluke > > calibration setup. To bad they tossed away all cables. > > -332B voltage standard, worked like a charm after changing one cap. > > -760A meter calibrator, worked for a week, now constant jumps in > > protecting and the home fuse blowes even if I put the plug in powered of > > so I think a Y capacitor is leak. I will find that, no problem (I hope) > > -510A AC reference standard, dead batteries but I just removed them > > because they are optional. I left in the option board. Still not sure if > > that is wise. > > - 731A, also dead nicads. I removed them and put in a 12V voltage > > regulator. Also had to replace a paper-foil capacitor. > > - 750A voltage divider. There is room for two strange batteries but as > > far as I understand it also workes without as long as you are sure You > > do not over voltage > > -some HP stuff like a 20 penta Ohm elctro Ohmmeter/voltage source, a > > time standard, pindiode modulator and low distrotion sinewave generator. > > > > But then two very important parts of the set up. A 720 kelvin Varley > > divider. All see-through plastic isolator standoffs and centre coupling > > parts of the switches are desintegrated. That will become a hell of a > > job. > > > > The biggest problem was my 845AB, they allready removed the nicads and > > added two extra 10V zeners. Thick metal multi-amp ones. The problem is > > the manual is very limited. There are some scope pictures and a small > > trouble seek section. > > I had to replace allmost every cap, so did them all. Then I kept > > problems with stability. Then I took my lab-supply and put 10V over the > > capacitor. It drew about 30mA and was stable. Then I raised the voltage > > and when the zener started conducting the 845 draw 600mA. I think the > > 10V mega-zeners started drawing so much current the transformer > > saturated or at least power got unstable. Could not close the top lid, > > as soon as it came near the transformer the meter went nuts. I now > > placed a 7812 and it is stable over the whole range I think. At 1uV full > > scale with shorted input and knob at zero the meter deviated plus/minus > > 0,1 uV max, most times about 0,05uV. I think that is not bad but this is > > my only null-detector so I do not know for sure. > > > > Now I have to find out how to make cables to connect everything, find a > > way to determine what will be my STANDARD VOLT. I have the calibrating > > history upto 1988. I can measure the differences between the cells but > > first have to find out the safest way, and if that is comform the old > > papers I think I take that as a basis. Then adjust the calibrators to > > that value. This will probably still be more accurate as the average > > multi or bench meter. I have a 6,5 digit as best one. > > > > Does some one has a better way ? > > > > > > For my collection See www.pa4tim.nl Not all the fluke stuff is on it > > yet. > > > > 73 > > > > Fred > > > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > and follow the instructions there. > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
W
WB6BNQ
Wed, Apr 27, 2011 3:34 AM

Hi Fred,

Well, it sounds like the people that had this equipment really had no idea how to care
for it properly.  It also suggests that they did not know what they were doing and just
bumbled along.  Could be the reason the operation was shut down at some point.

No special manuals exist for the 731, 845 or for that matter most of all of the old
Fluke equipment.  Any inside understanding was gained through experience.

The best method you have for comparing against the cells is to use the 750A divider.
This divider is very high quality and hopefully the previous owners did not go inside
it.  The only truly accurate and stable output on the 731 is the 10 volt posts.  I
would hook up the cells and the 845 to their respective connections on the 750A and
then attach the 731's 10 volt posts to the divider string's main input terminals (as
shown on page 2-5) with the switch set to 10 volts.  Then do your inter comparisons
with the different cells in the Guildline.  You can use the cell voltage level switches
on the 750A for determining the cell values with the 731 acting as the reference.

The best you will be able to do (unless our 731 is calibrated) is to determine
differences between cells.  By taking readings several times a day over many days and
weeks you can get a picture of the cell activity and their stability.

If your Fluke 731 is not calibrated, hopefully you may find a known quality Cal Lab
that would provide you access to adjust your 731 to a known value.  Then that would
become your main reference.  Then you could assign values to the cells in the
Guildline.  From that bootstrap you should be able to arrive at a confidence level to
trust what is happening with the Guildline and eventually make it your Lab reference.
Or at least make the Guildline a faithful sanity check.  The problem with cells is you
need to constantly inter compare them to make sure they are behaving correctly and
remaining stable.

As the equipment, presummably, has not been compared or adjusted against a real
standard, I am not surprised by the offset readings you measured.  The best stable
voltage reference you have is the Fluke 731.  It would be well worth the expense to
have it calibrated with a high quality Lab if you have one near you.

Regarding the 720, Fluke was making those until just very recently.  I would see if
they could still provide you with the necessary parts to rebuild it.  It would be well
worth the effort in my opinion.

73....Bill....WB6BNQ

Fred wrote:

I think I have the manuals complete, they are original and I do not miss
pages but maybe there is also a real extended service manual for the
845AB.
In my schematic there is a A and a AB powersupply schematic. The A has a
zener and no battery. They removed the battery in mine, that was also
written on the cabinet and placed two zeners. I think they did not get
it right anymore and then bought the Fluke 8500 that was part of the
setup (in the drawings I had in the documentation it was coupled to the
750)

Tonight I did some first measurement. The Cells are now about a month or
two operational and heated. They are very carefull moved to a special
made shelf two weeks ago. But now I have a problem. It's like the man
with just 1 clock, he think he does know the time, the man with two
clocks has a problem. I have 4 DC calibrators and 4 Weston cells so I'm
lost :-)

I first measured 1 cell from the Guildline using the 845 and the 731. I
turned the 731 dial until I had a zero in the 3uV range and allmost zero
in the 1uV range. I took that value as a reference. Following the trend
in the graphs from this cell (from 1977 upto 1987) the value could be
realistic. The oven temp is 30.07 degrees. The philips was about plus
250uV, the 332 was minus 160uV. But if I connect them through the 750 I
get a 25uV difference. I have not tried to calibrate it, it is always
kept in the same clean room and never messed around so it will be
probably better if I leave it like this because the 720 is not usefull
and needed for calibration.

The 720 is strange, it is clean as a baby inside-out. It is not been
dropped or so. But all the plastic is cracked and broken. If you touch
it it crumbles. Maybe someone back then cleaned it with some solvent.
It's a pitty because it is a beautifull instrument. I must find some
teflon or poly-ethilene to make new ones (I have a lath and a mill but
is is a lot of work) Some are possible repairable using 2K-epoxy. One
switch, number 1 from right, is very bad, I had to take it out because
the centre-follower is crumbled. The others still work and are in place
but the studs have cracks and pieces allready missing.

73

Fred

WB6BNQ schreef op di 26-04-2011 om 13:20 [-0700]:

Hi Fred,

Looks like you got lucky with such a nice collection being offered.

I am confused on why additional zeners would be added to the 731 and 845.  Both
were made to work without batteries.

The 731, if it is in good shape, is an extremely stable reference source.  Well
able to transfer a known value from a competent calibration Lab to use for
establishing the value of your Guildline cells.

As I am sure you know, the 845 is the only high impedance DC NULL meter available
that was worth a damn.  The first mains power transformer (and battery with its
charge circuit) is just to supply a power source for the 84 hertz transistor
inverter to the neon's.  That same inverter is also used to power the all of the
circuitry.  That is part of what provided the extremely high input isolation.
The noise you see on the 1 uv scale is normal.

The 720 switch problem sounds interesting, I wonder what caused the switches to
be so bad ?  In a clean environment there should be no reason for them to have
disintegrated.

The Guildline was a very nice catch.  Have you had an opportunity to determine if
the cells are in good shape ?  Hopefully nobody loaded them with a low
resistance, that would have messed them up for sure.  In conjunction with the
845, 750 and the 731 you have yourself a nice voltage reference system.

As for cables, well, I would not worry about that.  Just use normal copper wire
and you will not go wrong.

From your comments I am wondering if you have the full manuals for the 731 and
845 ?

73....Bill....WB6BNQ

Fred wrote:

Hello,

My name is fred, PA4TIM, I'm a volt-nut. Took a lot of therapy to
confess ;-)

I was already fascinated by calibrations because I restore measurement
equipment. I collected some old calibrators like an old one from Philips
using nixies as indicator, a Tek AC calibrator ment for internal factory
use.

But a few months ago a got a mail from a stranger who noticed my website
and asked if I was interested in the stuff from his companies old
calibrating lab (for free !!!) The lab was closed 25 years ago, but the
stuff stayed there unused. Now they needed the room and "found" this
stuff but almost nobody knew what it was so a lot was thrown away. He
thought that was a shame and asked if he was allowed to find a
collector. Al stuff including manuals and calibration history and in
state like new. But not used for 25 years and some stuff suffered from
sleeping or was maybe already death when parked away

It started with a Guildline 4 cell unit. In perefect condition. I had to
repair the heater and some minor aging things.

A tek 576 curve tracer came as a bonus with a box new Tek parts.

Then things speed up, if I wanted some more. A complete Fluke
calibration setup. To bad they tossed away all cables.
-332B voltage standard, worked like a charm after changing one cap.
-760A meter calibrator, worked for a week, now constant jumps in
protecting and the home fuse blowes even if I put the plug in powered of
so I think a Y capacitor is leak. I will find that, no problem (I hope)
-510A AC reference standard, dead batteries but I just removed them
because they are optional. I left in the option board. Still not sure if
that is wise.

  • 731A, also dead nicads. I removed them and put in a 12V voltage
    regulator. Also had to replace a paper-foil capacitor.
  • 750A voltage divider. There is room for two strange batteries but as
    far as I understand it also workes without as long as you are sure You
    do not over voltage
    -some HP stuff like a 20 penta Ohm elctro Ohmmeter/voltage source, a
    time standard, pindiode modulator and low distrotion sinewave generator.

But then two very important parts of the set up. A 720 kelvin Varley
divider. All see-through plastic isolator standoffs and centre coupling
parts of the switches are desintegrated. That will become a hell of a
job.

The biggest problem was my 845AB, they allready removed the nicads and
added two extra 10V zeners. Thick metal multi-amp ones.  The problem is
the manual is very limited. There are some scope pictures and a small
trouble seek section.
I had to replace allmost every cap, so did them all. Then I kept
problems with stability. Then I took my lab-supply and put 10V over the
capacitor. It drew about 30mA and was stable. Then I raised the voltage
and when the zener started conducting the 845 draw 600mA. I think the
10V mega-zeners started drawing so much current the transformer
saturated or at least power got unstable. Could not close the top lid,
as soon as it came near the transformer the meter went nuts. I now
placed a 7812 and it is stable over the whole range I think. At 1uV full
scale with shorted input and knob at zero the meter deviated plus/minus
0,1 uV max, most times about 0,05uV. I think that is not bad but this is
my only null-detector so I do not know for sure.

Now I have to find out how to make cables to connect everything, find a
way to determine what will be my  STANDARD VOLT. I have the calibrating
history upto 1988. I can measure the differences between the cells but
first have to find out the safest way, and if that is comform the old
papers I think I take that as a basis. Then adjust the calibrators to
that value. This will probably still be more accurate as the average
multi or bench meter. I have a 6,5 digit as best one.

Does some one has a better way ?

For my collection See www.pa4tim.nl Not all the fluke stuff is on it
yet.

73

Fred


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi Fred, Well, it sounds like the people that had this equipment really had no idea how to care for it properly. It also suggests that they did not know what they were doing and just bumbled along. Could be the reason the operation was shut down at some point. No special manuals exist for the 731, 845 or for that matter most of all of the old Fluke equipment. Any inside understanding was gained through experience. The best method you have for comparing against the cells is to use the 750A divider. This divider is very high quality and hopefully the previous owners did not go inside it. The only truly accurate and stable output on the 731 is the 10 volt posts. I would hook up the cells and the 845 to their respective connections on the 750A and then attach the 731's 10 volt posts to the divider string's main input terminals (as shown on page 2-5) with the switch set to 10 volts. Then do your inter comparisons with the different cells in the Guildline. You can use the cell voltage level switches on the 750A for determining the cell values with the 731 acting as the reference. The best you will be able to do (unless our 731 is calibrated) is to determine differences between cells. By taking readings several times a day over many days and weeks you can get a picture of the cell activity and their stability. If your Fluke 731 is not calibrated, hopefully you may find a known quality Cal Lab that would provide you access to adjust your 731 to a known value. Then that would become your main reference. Then you could assign values to the cells in the Guildline. From that bootstrap you should be able to arrive at a confidence level to trust what is happening with the Guildline and eventually make it your Lab reference. Or at least make the Guildline a faithful sanity check. The problem with cells is you need to constantly inter compare them to make sure they are behaving correctly and remaining stable. As the equipment, presummably, has not been compared or adjusted against a real standard, I am not surprised by the offset readings you measured. The best stable voltage reference you have is the Fluke 731. It would be well worth the expense to have it calibrated with a high quality Lab if you have one near you. Regarding the 720, Fluke was making those until just very recently. I would see if they could still provide you with the necessary parts to rebuild it. It would be well worth the effort in my opinion. 73....Bill....WB6BNQ Fred wrote: > I think I have the manuals complete, they are original and I do not miss > pages but maybe there is also a real extended service manual for the > 845AB. > In my schematic there is a A and a AB powersupply schematic. The A has a > zener and no battery. They removed the battery in mine, that was also > written on the cabinet and placed two zeners. I think they did not get > it right anymore and then bought the Fluke 8500 that was part of the > setup (in the drawings I had in the documentation it was coupled to the > 750) > > Tonight I did some first measurement. The Cells are now about a month or > two operational and heated. They are very carefull moved to a special > made shelf two weeks ago. But now I have a problem. It's like the man > with just 1 clock, he think he does know the time, the man with two > clocks has a problem. I have 4 DC calibrators and 4 Weston cells so I'm > lost :-) > > I first measured 1 cell from the Guildline using the 845 and the 731. I > turned the 731 dial until I had a zero in the 3uV range and allmost zero > in the 1uV range. I took that value as a reference. Following the trend > in the graphs from this cell (from 1977 upto 1987) the value could be > realistic. The oven temp is 30.07 degrees. The philips was about plus > 250uV, the 332 was minus 160uV. But if I connect them through the 750 I > get a 25uV difference. I have not tried to calibrate it, it is always > kept in the same clean room and never messed around so it will be > probably better if I leave it like this because the 720 is not usefull > and needed for calibration. > > The 720 is strange, it is clean as a baby inside-out. It is not been > dropped or so. But all the plastic is cracked and broken. If you touch > it it crumbles. Maybe someone back then cleaned it with some solvent. > It's a pitty because it is a beautifull instrument. I must find some > teflon or poly-ethilene to make new ones (I have a lath and a mill but > is is a lot of work) Some are possible repairable using 2K-epoxy. One > switch, number 1 from right, is very bad, I had to take it out because > the centre-follower is crumbled. The others still work and are in place > but the studs have cracks and pieces allready missing. > > 73 > > Fred > > WB6BNQ schreef op di 26-04-2011 om 13:20 [-0700]: > > Hi Fred, > > > > Looks like you got lucky with such a nice collection being offered. > > > > I am confused on why additional zeners would be added to the 731 and 845. Both > > were made to work without batteries. > > > > The 731, if it is in good shape, is an extremely stable reference source. Well > > able to transfer a known value from a competent calibration Lab to use for > > establishing the value of your Guildline cells. > > > > As I am sure you know, the 845 is the only high impedance DC NULL meter available > > that was worth a damn. The first mains power transformer (and battery with its > > charge circuit) is just to supply a power source for the 84 hertz transistor > > inverter to the neon's. That same inverter is also used to power the all of the > > circuitry. That is part of what provided the extremely high input isolation. > > The noise you see on the 1 uv scale is normal. > > > > The 720 switch problem sounds interesting, I wonder what caused the switches to > > be so bad ? In a clean environment there should be no reason for them to have > > disintegrated. > > > > The Guildline was a very nice catch. Have you had an opportunity to determine if > > the cells are in good shape ? Hopefully nobody loaded them with a low > > resistance, that would have messed them up for sure. In conjunction with the > > 845, 750 and the 731 you have yourself a nice voltage reference system. > > > > As for cables, well, I would not worry about that. Just use normal copper wire > > and you will not go wrong. > > > > From your comments I am wondering if you have the full manuals for the 731 and > > 845 ? > > > > 73....Bill....WB6BNQ > > > > > > Fred wrote: > > > > > Hello, > > > > > > My name is fred, PA4TIM, I'm a volt-nut. Took a lot of therapy to > > > confess ;-) > > > > > > I was already fascinated by calibrations because I restore measurement > > > equipment. I collected some old calibrators like an old one from Philips > > > using nixies as indicator, a Tek AC calibrator ment for internal factory > > > use. > > > > > > But a few months ago a got a mail from a stranger who noticed my website > > > and asked if I was interested in the stuff from his companies old > > > calibrating lab (for free !!!) The lab was closed 25 years ago, but the > > > stuff stayed there unused. Now they needed the room and "found" this > > > stuff but almost nobody knew what it was so a lot was thrown away. He > > > thought that was a shame and asked if he was allowed to find a > > > collector. Al stuff including manuals and calibration history and in > > > state like new. But not used for 25 years and some stuff suffered from > > > sleeping or was maybe already death when parked away > > > > > > It started with a Guildline 4 cell unit. In perefect condition. I had to > > > repair the heater and some minor aging things. > > > > > > A tek 576 curve tracer came as a bonus with a box new Tek parts. > > > > > > Then things speed up, if I wanted some more. A complete Fluke > > > calibration setup. To bad they tossed away all cables. > > > -332B voltage standard, worked like a charm after changing one cap. > > > -760A meter calibrator, worked for a week, now constant jumps in > > > protecting and the home fuse blowes even if I put the plug in powered of > > > so I think a Y capacitor is leak. I will find that, no problem (I hope) > > > -510A AC reference standard, dead batteries but I just removed them > > > because they are optional. I left in the option board. Still not sure if > > > that is wise. > > > - 731A, also dead nicads. I removed them and put in a 12V voltage > > > regulator. Also had to replace a paper-foil capacitor. > > > - 750A voltage divider. There is room for two strange batteries but as > > > far as I understand it also workes without as long as you are sure You > > > do not over voltage > > > -some HP stuff like a 20 penta Ohm elctro Ohmmeter/voltage source, a > > > time standard, pindiode modulator and low distrotion sinewave generator. > > > > > > But then two very important parts of the set up. A 720 kelvin Varley > > > divider. All see-through plastic isolator standoffs and centre coupling > > > parts of the switches are desintegrated. That will become a hell of a > > > job. > > > > > > The biggest problem was my 845AB, they allready removed the nicads and > > > added two extra 10V zeners. Thick metal multi-amp ones. The problem is > > > the manual is very limited. There are some scope pictures and a small > > > trouble seek section. > > > I had to replace allmost every cap, so did them all. Then I kept > > > problems with stability. Then I took my lab-supply and put 10V over the > > > capacitor. It drew about 30mA and was stable. Then I raised the voltage > > > and when the zener started conducting the 845 draw 600mA. I think the > > > 10V mega-zeners started drawing so much current the transformer > > > saturated or at least power got unstable. Could not close the top lid, > > > as soon as it came near the transformer the meter went nuts. I now > > > placed a 7812 and it is stable over the whole range I think. At 1uV full > > > scale with shorted input and knob at zero the meter deviated plus/minus > > > 0,1 uV max, most times about 0,05uV. I think that is not bad but this is > > > my only null-detector so I do not know for sure. > > > > > > Now I have to find out how to make cables to connect everything, find a > > > way to determine what will be my STANDARD VOLT. I have the calibrating > > > history upto 1988. I can measure the differences between the cells but > > > first have to find out the safest way, and if that is comform the old > > > papers I think I take that as a basis. Then adjust the calibrators to > > > that value. This will probably still be more accurate as the average > > > multi or bench meter. I have a 6,5 digit as best one. > > > > > > Does some one has a better way ? > > > > > > > > > For my collection See www.pa4tim.nl Not all the fluke stuff is on it > > > yet. > > > > > > 73 > > > > > > Fred > > > > > > _______________________________________________ > > > volt-nuts mailing list -- volt-nuts@febo.com > > > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > > and follow the instructions there. > > > > > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > and follow the instructions there. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
PK
Poul-Henning Kamp
Wed, Apr 27, 2011 5:37 AM

In message 1303856302.1630.51.camel@Fred-asustower, Fred writes:

The 720 is strange, it is clean as a baby inside-out. It is not been
dropped or so. But all the plastic is cracked and broken. If you touch
it it crumbles.

We see this kinde of thing happen in computer museums too, it's a
production fault: It took some time until the plastic industry found
long term stable additives.

This is also why serious museums worry a lot about their barbie
doll collections.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

In message <1303856302.1630.51.camel@Fred-asustower>, Fred writes: >The 720 is strange, it is clean as a baby inside-out. It is not been >dropped or so. But all the plastic is cracked and broken. If you touch >it it crumbles. We see this kinde of thing happen in computer museums too, it's a production fault: It took some time until the plastic industry found long term stable additives. This is also why serious museums worry a lot about their barbie doll collections. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.
ME
Marvin E. Gozum
Thu, Apr 28, 2011 1:34 PM

Is there a way to decipher plastics potential longevity in a finished
product?

I'm impressed plastic parts on many HP equipment endure intact over
20+ years, some barely discoloring.  I can't tell what type they are,
but the more durable plastics often feel hefty and solid.  Likewise,
Fluke DMM have the same 'feeling' and endure for decades.

On a side note, casing of some iPhones barely 2-3 years old are
spontaneously cracking, so it can be made quite badly too.

At 01:37 AM 4/27/2011, Poul-Henning Kamp wrote:

In message 1303856302.1630.51.camel@Fred-asustower, Fred writes:

The 720 is strange, it is clean as a baby inside-out. It is not been
dropped or so. But all the plastic is cracked and broken. If you touch
it it crumbles.

We see this kinde of thing happen in computer museums too, it's a
production fault: It took some time until the plastic industry found
long term stable additives.

This is also why serious museums worry a lot about their barbie
doll collections.

Sincerely,

Marv Gozum
Philadelphia, PA

Is there a way to decipher plastics potential longevity in a finished product? I'm impressed plastic parts on many HP equipment endure intact over 20+ years, some barely discoloring. I can't tell what type they are, but the more durable plastics often feel hefty and solid. Likewise, Fluke DMM have the same 'feeling' and endure for decades. On a side note, casing of some iPhones barely 2-3 years old are spontaneously cracking, so it can be made quite badly too. At 01:37 AM 4/27/2011, Poul-Henning Kamp wrote: >In message <1303856302.1630.51.camel@Fred-asustower>, Fred writes: > > >The 720 is strange, it is clean as a baby inside-out. It is not been > >dropped or so. But all the plastic is cracked and broken. If you touch > >it it crumbles. > >We see this kinde of thing happen in computer museums too, it's a >production fault: It took some time until the plastic industry found >long term stable additives. > >This is also why serious museums worry a lot about their barbie >doll collections. Sincerely, Marv Gozum Philadelphia, PA
CH
Chuck Harris
Thu, Apr 28, 2011 2:36 PM

The only real way to tell with plastics is to subject them to the test
of time and environment.  The basic problem is one of plasticizer migration.
Many plastics have an added goo that improves the way they flow in molding,
and softens the plastic so that it is suitable for use.  The plasticizer
evaporates slowly out of the plastic making it shrink, and leaving it very
brittle.  Some plasticizers migrate to the surface leaving the plastic very
sticky... vinyl (PVC), particularly soft vinyl, which is heavy in plasticizer,
has that problem.  That plasticizer migration is what makes xerox'd pages
stick to vinyl binders, leaving the lettering behind on the vinyl...

Some of the ancient plastics aren't really plastic at all... bakelite is an
example.... and they hold up incredibly well.  The common characteristic,
in my opinion, no plasticizer.  The much fabled BPA is one of the plasticizers
that causes problems if it isn't used just right.... It also causes breast
development in men, but that is another story.

-Chuck Harris

Marvin E. Gozum wrote:

Is there a way to decipher plastics potential longevity in a finished
product?

I'm impressed plastic parts on many HP equipment endure intact over 20+
years, some barely discoloring. I can't tell what type they are, but the
more durable plastics often feel hefty and solid. Likewise, Fluke DMM
have the same 'feeling' and endure for decades.

On a side note, casing of some iPhones barely 2-3 years old are
spontaneously cracking, so it can be made quite badly too.

The only real way to tell with plastics is to subject them to the test of time and environment. The basic problem is one of plasticizer migration. Many plastics have an added goo that improves the way they flow in molding, and softens the plastic so that it is suitable for use. The plasticizer evaporates slowly out of the plastic making it shrink, and leaving it very brittle. Some plasticizers migrate to the surface leaving the plastic very sticky... vinyl (PVC), particularly soft vinyl, which is heavy in plasticizer, has that problem. That plasticizer migration is what makes xerox'd pages stick to vinyl binders, leaving the lettering behind on the vinyl... Some of the ancient plastics aren't really plastic at all... bakelite is an example.... and they hold up incredibly well. The common characteristic, in my opinion, no plasticizer. The much fabled BPA is one of the plasticizers that causes problems if it isn't used just right.... It also causes breast development in men, but that is another story. -Chuck Harris Marvin E. Gozum wrote: > Is there a way to decipher plastics potential longevity in a finished > product? > > I'm impressed plastic parts on many HP equipment endure intact over 20+ > years, some barely discoloring. I can't tell what type they are, but the > more durable plastics often feel hefty and solid. Likewise, Fluke DMM > have the same 'feeling' and endure for decades. > > On a side note, casing of some iPhones barely 2-3 years old are > spontaneously cracking, so it can be made quite badly too.
F
Fred
Thu, Apr 28, 2011 5:03 PM

I have removed one switch. I have measured the resistance at 500V from
the (cleaned) original material and that was much lower as a same lenght
of teflon. (but the Teflon was very hard to measure, if I came closer
then a meter to the HP4329A the meter jumped around, the needle was
almost max left in the most upper range at 1000V so we are talking
penta-ohms, I'm not only a volt-nut ;-) ).

So I made out of teflon a shaft-to-rotor coupler. A hell of a job. took
me two hours  milling and turning the lath but then it fitted like a
glove. Then I made 12 teflon studs. I drilled a 4mm hole and with a
little tool I made I could press in the old inserts. The switch is
already in its place.

But I noticed a strange thing. The stud that was broken in to pieces was
hard and brittle. One insert was still in but just a little hit with a
screwdriver and it broke in two.
The other stud was still in one piece. But I did not want to take a risk
so i thought, lets replace this one too. I gave it a gentle hit but
nothing, took a pair of pliers squeezed as hard as I could, nothing.
even with a cutter, it just left bitemarks but no breaking. After all
this violence I had to replace it but I think there was nothing wrong
with this one. I heated it and then I could pull the inserts out.

So this is weird, almost all studs are gone to dust but some are as
strong as new. The problem is, to test it I have to use force and that
weakens them so I exchange them all to be shure.

But making 5 extra new rotors is a hell of a job so I will check the
rotor couplers optical very careful for cracks and leave them. The
reason the studs break when the plastic get hard is because the inserts
are a bit conical so screwing in the bold expands them. So there is
tension in the material. The rotor couplers are just a "lose" fit and if
one breaks I now know how to replace them. Not much work. The broken
coupler was coupled to a floating one rotorswitch  (both studs complete
lose) so maybe a shock during transport.

I also mailed Fluke here in the Netherlands. Got a mail back from the
service desk they forwarded it to the service manager. If they can
supply new rotors and studs (affordable)  that is even better, I like it
to be original.

Now I am going to find out if there is a affortable solution to
calibrate and adjust the 731A. I spoke someone who had a Tek current
scope probe calibrated and adjusted, That was allmost 500 euro. To much
for me.
After that I can compare them with my Guildline. (if I remember well the
10V is not adjustable but I have the manual so I will look that up
first)

Fred

Chuck Harris schreef op do 28-04-2011 om 10:36 [-0400]:

The only real way to tell with plastics is to subject them to the test
of time and environment.  The basic problem is one of plasticizer migration.
Many plastics have an added goo that improves the way they flow in molding,
and softens the plastic so that it is suitable for use.  The plasticizer
evaporates slowly out of the plastic making it shrink, and leaving it very
brittle.  Some plasticizers migrate to the surface leaving the plastic very
sticky... vinyl (PVC), particularly soft vinyl, which is heavy in plasticizer,
has that problem.  That plasticizer migration is what makes xerox'd pages
stick to vinyl binders, leaving the lettering behind on the vinyl...

Some of the ancient plastics aren't really plastic at all... bakelite is an
example.... and they hold up incredibly well.  The common characteristic,
in my opinion, no plasticizer.  The much fabled BPA is one of the plasticizers
that causes problems if it isn't used just right.... It also causes breast
development in men, but that is another story.

-Chuck Harris

Marvin E. Gozum wrote:

Is there a way to decipher plastics potential longevity in a finished
product?

I'm impressed plastic parts on many HP equipment endure intact over 20+
years, some barely discoloring. I can't tell what type they are, but the
more durable plastics often feel hefty and solid. Likewise, Fluke DMM
have the same 'feeling' and endure for decades.

On a side note, casing of some iPhones barely 2-3 years old are
spontaneously cracking, so it can be made quite badly too.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

I have removed one switch. I have measured the resistance at 500V from the (cleaned) original material and that was much lower as a same lenght of teflon. (but the Teflon was very hard to measure, if I came closer then a meter to the HP4329A the meter jumped around, the needle was almost max left in the most upper range at 1000V so we are talking penta-ohms, I'm not only a volt-nut ;-) ). So I made out of teflon a shaft-to-rotor coupler. A hell of a job. took me two hours milling and turning the lath but then it fitted like a glove. Then I made 12 teflon studs. I drilled a 4mm hole and with a little tool I made I could press in the old inserts. The switch is already in its place. But I noticed a strange thing. The stud that was broken in to pieces was hard and brittle. One insert was still in but just a little hit with a screwdriver and it broke in two. The other stud was still in one piece. But I did not want to take a risk so i thought, lets replace this one too. I gave it a gentle hit but nothing, took a pair of pliers squeezed as hard as I could, nothing. even with a cutter, it just left bitemarks but no breaking. After all this violence I had to replace it but I think there was nothing wrong with this one. I heated it and then I could pull the inserts out. So this is weird, almost all studs are gone to dust but some are as strong as new. The problem is, to test it I have to use force and that weakens them so I exchange them all to be shure. But making 5 extra new rotors is a hell of a job so I will check the rotor couplers optical very careful for cracks and leave them. The reason the studs break when the plastic get hard is because the inserts are a bit conical so screwing in the bold expands them. So there is tension in the material. The rotor couplers are just a "lose" fit and if one breaks I now know how to replace them. Not much work. The broken coupler was coupled to a floating one rotorswitch (both studs complete lose) so maybe a shock during transport. I also mailed Fluke here in the Netherlands. Got a mail back from the service desk they forwarded it to the service manager. If they can supply new rotors and studs (affordable) that is even better, I like it to be original. Now I am going to find out if there is a affortable solution to calibrate and adjust the 731A. I spoke someone who had a Tek current scope probe calibrated and adjusted, That was allmost 500 euro. To much for me. After that I can compare them with my Guildline. (if I remember well the 10V is not adjustable but I have the manual so I will look that up first) Fred Chuck Harris schreef op do 28-04-2011 om 10:36 [-0400]: > The only real way to tell with plastics is to subject them to the test > of time and environment. The basic problem is one of plasticizer migration. > Many plastics have an added goo that improves the way they flow in molding, > and softens the plastic so that it is suitable for use. The plasticizer > evaporates slowly out of the plastic making it shrink, and leaving it very > brittle. Some plasticizers migrate to the surface leaving the plastic very > sticky... vinyl (PVC), particularly soft vinyl, which is heavy in plasticizer, > has that problem. That plasticizer migration is what makes xerox'd pages > stick to vinyl binders, leaving the lettering behind on the vinyl... > > Some of the ancient plastics aren't really plastic at all... bakelite is an > example.... and they hold up incredibly well. The common characteristic, > in my opinion, no plasticizer. The much fabled BPA is one of the plasticizers > that causes problems if it isn't used just right.... It also causes breast > development in men, but that is another story. > > -Chuck Harris > > Marvin E. Gozum wrote: > > Is there a way to decipher plastics potential longevity in a finished > > product? > > > > I'm impressed plastic parts on many HP equipment endure intact over 20+ > > years, some barely discoloring. I can't tell what type they are, but the > > more durable plastics often feel hefty and solid. Likewise, Fluke DMM > > have the same 'feeling' and endure for decades. > > > > On a side note, casing of some iPhones barely 2-3 years old are > > spontaneously cracking, so it can be made quite badly too. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
MG
Marv Gozum @ JHN
Thu, Apr 28, 2011 7:04 PM

Thanks Chuck, that's very informative.  That pretty much describes
the life cycle of a lot of old plastic I've encountered.

You say evaporate, is possible then to seal plastic parts to prevent
plasticizer migration, and prolong the parts life, even if its
already old but still has not fallen apart?  Add any kind of
plasticizer to revitalize old plastic?

Yes, vinyl binders seem to almost weep something, that doesn't seem
good!  I see Bakelite endures over time, but it very ceramic like and
tends to cracks if dropped.

Sorry all, at most this should be discussed at the HP forum group, so
this is my last topic digression on the volt-nuts forum.

At 10:36 AM 4/28/2011, Chuck Harris wrote:

The only real way to tell with plastics is to subject them to the test
of time and environment.  The basic problem is one of plasticizer migration.
Many plastics have an added goo that improves the way they flow in molding,
and softens the plastic so that it is suitable for use.  The plasticizer
evaporates slowly out of the plastic making it shrink, and leaving it very
brittle.  Some plasticizers migrate to the surface leaving the plastic very
sticky... vinyl (PVC), particularly soft vinyl, which is heavy in plasticizer,
has that problem.  That plasticizer migration is what makes xerox'd pages
stick to vinyl binders, leaving the lettering behind on the vinyl...

Some of the ancient plastics aren't really plastic at all... bakelite is an
example.... and they hold up incredibly well.  The common characteristic,
in my opinion, no plasticizer.  The much fabled BPA is one of the plasticizers
that causes problems if it isn't used just right.... It also causes breast
development in men, but that is another story.

-Chuck Harris

Best Wishes,

Marv Gozum
Philadelphia

Thanks Chuck, that's very informative. That pretty much describes the life cycle of a lot of old plastic I've encountered. You say evaporate, is possible then to seal plastic parts to prevent plasticizer migration, and prolong the parts life, even if its already old but still has not fallen apart? Add any kind of plasticizer to revitalize old plastic? Yes, vinyl binders seem to almost weep something, that doesn't seem good! I see Bakelite endures over time, but it very ceramic like and tends to cracks if dropped. Sorry all, at most this should be discussed at the HP forum group, so this is my last topic digression on the volt-nuts forum. At 10:36 AM 4/28/2011, Chuck Harris wrote: >The only real way to tell with plastics is to subject them to the test >of time and environment. The basic problem is one of plasticizer migration. >Many plastics have an added goo that improves the way they flow in molding, >and softens the plastic so that it is suitable for use. The plasticizer >evaporates slowly out of the plastic making it shrink, and leaving it very >brittle. Some plasticizers migrate to the surface leaving the plastic very >sticky... vinyl (PVC), particularly soft vinyl, which is heavy in plasticizer, >has that problem. That plasticizer migration is what makes xerox'd pages >stick to vinyl binders, leaving the lettering behind on the vinyl... > >Some of the ancient plastics aren't really plastic at all... bakelite is an >example.... and they hold up incredibly well. The common characteristic, >in my opinion, no plasticizer. The much fabled BPA is one of the plasticizers >that causes problems if it isn't used just right.... It also causes breast >development in men, but that is another story. > >-Chuck Harris Best Wishes, Marv Gozum Philadelphia