AJ
Andreas Jahn
Fri, Nov 11, 2011 5:33 AM
Hello Randy,
with a capacitive divider 2:1 or 3:1 based on a LTC1043 + LT1051
you can extend the 5V range to 10 or 15 V. The capacitive divider
is very stable over temperature and time.
With my 2:1 dividers I have less than 1 ppm drift over a
10-40 degrees (celsius) temperature range.
Absolute accuracy of the divider factor depends mainly on
charge injection of the switch and capacitor losses.
I measured values between 2-10ppm depending on different
pin configurations. Im not shure wether the examples in the
datasheet are using the correct pin numbers for minimum charge injection.
Another Question: how do you get a unipolar 0..5V range with a ADS1282?
The absolute input of the PGA-Amplifier is specified from 0.7 .. 3.75V.
With best regards
Andreas
----- Original Message -----
From: "Randy Evans" randallgrayevans@yahoo.com
To: "Михаил" timka2k@yandex.ru; "Discussion of precise voltage
measurement" volt-nuts@febo.com
Sent: Friday, November 11, 2011 5:01 AM
Subject: Re: [volt-nuts] Homebrew DVM
The ADS1282 alswo only has a 5V unipolar input range and I want a 0 to 12V
input range without dividers. I don't have any information on the
Solartron 7081.
Randy
From: Михаил timka2k@yandex.ru
To: Randy Evans randallgrayevans@yahoo.com; Discussion of precise
voltage measurement volt-nuts@febo.com
Sent: Thursday, November 10, 2011 9:17 AM
Subject: Re[2]: [volt-nuts] Homebrew DVM
Hi, Randy!
AN-260 is a very bad example of slow medium scale ADC.
I think that 130 db SNR ADS1282 is much better. Also, 8,5-digit Solartron
7081
have a simple and cheap delta-sigma ADC with 0.06 ppm noise and 51 sec.
integration time.
Mickle T.
Thursday, November 10, 2011, 8:20:58 PM, you wrote:
RE> 8.5 digits with averaging.
RE>
RE> I did consider that but the ADCs I am familiar with only have a 5V
input
RE> range so range resistors would be required, not desirable for me.
Also,
RE> they won't directly take the 7.2V ref input from a LTZ1000. Plus the
AN-260
RE> circuitry is not too complex. Otherwise, I agree with you.
RE>
RE> Randy
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Hello Randy,
with a capacitive divider 2:1 or 3:1 based on a LTC1043 + LT1051
you can extend the 5V range to 10 or 15 V. The capacitive divider
is very stable over temperature and time.
With my 2:1 dividers I have less than 1 ppm drift over a
10-40 degrees (celsius) temperature range.
Absolute accuracy of the divider factor depends mainly on
charge injection of the switch and capacitor losses.
I measured values between 2-10ppm depending on different
pin configurations. Im not shure wether the examples in the
datasheet are using the correct pin numbers for minimum charge injection.
Another Question: how do you get a unipolar 0..5V range with a ADS1282?
The absolute input of the PGA-Amplifier is specified from 0.7 .. 3.75V.
With best regards
Andreas
----- Original Message -----
From: "Randy Evans" <randallgrayevans@yahoo.com>
To: "Михаил" <timka2k@yandex.ru>; "Discussion of precise voltage
measurement" <volt-nuts@febo.com>
Sent: Friday, November 11, 2011 5:01 AM
Subject: Re: [volt-nuts] Homebrew DVM
> The ADS1282 alswo only has a 5V unipolar input range and I want a 0 to 12V
> input range without dividers. I don't have any information on the
> Solartron 7081.
>
> Randy
>
>
> ________________________________
> From: Михаил <timka2k@yandex.ru>
> To: Randy Evans <randallgrayevans@yahoo.com>; Discussion of precise
> voltage measurement <volt-nuts@febo.com>
> Sent: Thursday, November 10, 2011 9:17 AM
> Subject: Re[2]: [volt-nuts] Homebrew DVM
>
> Hi, Randy!
> AN-260 is a very bad example of slow medium scale ADC.
> I think that 130 db SNR ADS1282 is much better. Also, 8,5-digit Solartron
> 7081
> have a simple and cheap delta-sigma ADC with 0.06 ppm noise and 51 sec.
> integration time.
>
>
> Mickle T.
>
> Thursday, November 10, 2011, 8:20:58 PM, you wrote:
> RE> 8.5 digits with averaging.
> RE>
> RE> I did consider that but the ADCs I am familiar with only have a 5V
> input
> RE> range so range resistors would be required, not desirable for me.
> Also,
> RE> they won't directly take the 7.2V ref input from a LTZ1000. Plus the
> AN-260
> RE> circuitry is not too complex. Otherwise, I agree with you.
> RE>
> RE> Randy
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
MK
m k
Fri, Nov 11, 2011 7:28 AM
Hi,
I have not built one yet, but it seems to be one of a very few that goes beyond 24 bits.
MK
From: Andreas_-_Jahn@t-online.de
To: volt-nuts@febo.com
Date: Thu, 10 Nov 2011 23:48:10 +0100
Subject: Re: [volt-nuts] Homebrew DVM
Hello MK,
did you already build a real cirquit with the LTC2440?
It has a differential input 2.5+/-2.5V for bridge measurement cirquits like
weight scales.
So I think it will be difficult to get a unipolar 0..5V measurement out of
this device without resistive divider.
Which buffer cirquits do you use to feed the low-impedant (capacitive)
inputs?
With best regards
Andreas
----- Original Message -----
From: "m k" m1k3k1@hotmail.com
To: volt-nuts@febo.com
Sent: Thursday, November 10, 2011 10:50 PM
Subject: Re: [volt-nuts] Homebrew DVM
Hi Rob,
It was the LTC 2440 and 2442.
Ehm, a search on LTC's website for "LTC2441" returns this:
Your search - ltc2441 - did not match any documents.
No pages were found containing "ltc2441".
No 8.5Hz, 26-bit converter shows up in the parametric search, either.
Hi,
I have not built one yet, but it seems to be one of a very few that goes beyond 24 bits.
MK
> From: Andreas_-_Jahn@t-online.de
> To: volt-nuts@febo.com
> Date: Thu, 10 Nov 2011 23:48:10 +0100
> Subject: Re: [volt-nuts] Homebrew DVM
>
> Hello MK,
>
> did you already build a real cirquit with the LTC2440?
> It has a differential input 2.5+/-2.5V for bridge measurement cirquits like
> weight scales.
> So I think it will be difficult to get a unipolar 0..5V measurement out of
> this device without resistive divider.
>
> Which buffer cirquits do you use to feed the low-impedant (capacitive)
> inputs?
>
> With best regards
>
> Andreas
>
> ----- Original Message -----
> From: "m k" <m1k3k1@hotmail.com>
> To: <volt-nuts@febo.com>
> Sent: Thursday, November 10, 2011 10:50 PM
> Subject: Re: [volt-nuts] Homebrew DVM
>
>
> >
> > Hi Rob,
> > It was the LTC 2440 and 2442.
> >
> >> Date: Thu, 10 Nov 2011 22:36:10 +0100
> >> From: rob.klein@smalldesign.nl
> >> To: volt-nuts@febo.com
> >> Subject: Re: [volt-nuts] Homebrew DVM
> >>
> >> Op 10-11-2011 20:39, m k schreef:
> >> > Hi Rob, Try the LTC2441
> >>
> >> Ehm, a search on LTC's website for "LTC2441" returns this:
> >>
> >> Your search - *ltc2441* - did not match any documents.
> >> No pages were found containing *"ltc2441"*.
> >>
> >> No 8.5Hz, 26-bit converter shows up in the parametric search, either.
> >>
> >>
> >> - Rob.
> >>
JD
John Devereux
Mon, Nov 14, 2011 7:42 AM
I am in the process of designing a homebrew DVM that initially will
only cover 0 to 12VDC measurement range. Perhaps in the future it
could be expended for a wider range but the issue is getting very
accurate and stable (stable is the key operative issue) range divider
resistors. The design is based on the National Semiconductor app note
260, but modified with more modern components such as low charge
injection analog switches, faster comparators, LTZ1000 reference,
extremely linear integrator, low cost uprocessor, etc. My hope is
that the DVM will be able to measure to 8.5 digits over this limited
range. We shall see. It should be a fun project, regardless.
The big issue is the lack of noise rejection characteristics. A
substantial amount of averaging may be required so fast 8.5 digit
measurements might be out of the question. However, it should be a
low cost measurement technique for high accuracy measurements of
stable voltages. Obviously, the measurement accuracy would be largely
determined by the accuracy of the LTZ1000 reference.
Since the integrator is the key circuit for this design, I would
really like to see the integrator design in the HP-3458A, if anyone
could provide a copy of the circuit. Comments on the approach would
be apprecitated.
Hi Randy,
I have an ambition to do this too. But I think there are better circuit
arrangements that are not really any more complicated. Or at least it
seems that way right now, before I have tried to make them work!
Firstly you have to make a pretty good circuit before you can beat a
cheap single-chip sigma-delta ADC, as others have pointed out. That
application note predates these I think.
You could look at some of the patents on multislope converters (as used
in the HP3458A, and all the top meters as far as I know). They are
surprisingly readable. Also many of them have expired by now, although
that does not matter for educational use.
The "basic" HP patent is 4357600:
http://www.google.com/patents?id=vCUxAAAAEBAJ
I think this is the HP3458A.
There is a later HP one 4951053 which refines it with a greatly improved
switching technique http://www.google.com/patents?id=INImAAAAEBAJ. But
the 3458A does not use it AFAIK.
Then fluke get in the game with 5321403 and 5565869, describing a
switching sequence with imnproved elimination of charge injection
errors. (These have not expired).
There are lots of others if you search on "multislope" but I found these
the most interesting.
--
John Devereux
Randy Evans <randallgrayevans@yahoo.com> writes:
> I am in the process of designing a homebrew DVM that initially will
> only cover 0 to 12VDC measurement range. Perhaps in the future it
> could be expended for a wider range but the issue is getting very
> accurate and stable (stable is the key operative issue) range divider
> resistors. The design is based on the National Semiconductor app note
> 260, but modified with more modern components such as low charge
> injection analog switches, faster comparators, LTZ1000 reference,
> extremely linear integrator, low cost uprocessor, etc. My hope is
> that the DVM will be able to measure to 8.5 digits over this limited
> range. We shall see. It should be a fun project, regardless.
> The big issue is the lack of noise rejection characteristics. A
> substantial amount of averaging may be required so fast 8.5 digit
> measurements might be out of the question. However, it should be a
> low cost measurement technique for high accuracy measurements of
> stable voltages. Obviously, the measurement accuracy would be largely
> determined by the accuracy of the LTZ1000 reference.
> Since the integrator is the key circuit for this design, I would
> really like to see the integrator design in the HP-3458A, if anyone
> could provide a copy of the circuit. Comments on the approach would
> be apprecitated.
Hi Randy,
I have an ambition to do this too. But I think there are better circuit
arrangements that are not really any more complicated. Or at least it
seems that way right now, before I have tried to make them work!
Firstly you have to make a pretty good circuit before you can beat a
cheap single-chip sigma-delta ADC, as others have pointed out. That
application note predates these I think.
You could look at some of the patents on multislope converters (as used
in the HP3458A, and all the top meters as far as I know). They are
surprisingly readable. Also many of them have expired by now, although
that does not matter for educational use.
The "basic" HP patent is 4357600:
<http://www.google.com/patents?id=vCUxAAAAEBAJ>
I think this is the HP3458A.
There is a later HP one 4951053 which refines it with a greatly improved
switching technique <http://www.google.com/patents?id=INImAAAAEBAJ>. But
the 3458A does not use it AFAIK.
Then fluke get in the game with 5321403 and 5565869, describing a
switching sequence with imnproved elimination of charge injection
errors. (These have not expired).
There are lots of others if you search on "multislope" but I found these
the most interesting.
--
John Devereux
RE
Randy Evans
Mon, Nov 14, 2011 11:10 PM
John,
Thanks for the info. I will read them tonight and assess wether I want to proceed with my original plan or revisit it. I have no particular preference, just what is easiest and cheapest but high resolution and accurate.
Randy
From: John Devereux john@devereux.me.uk
To: volt-nuts@febo.com
Cc: Discussion of precise voltage measurement volt-nuts@febo.com
Sent: Sunday, November 13, 2011 11:42 PM
Subject: Re: [volt-nuts] Homebrew DVM
Randy Evans randallgrayevans@yahoo.com writes:
I am in the process of designing a homebrew DVM that initially will
only cover 0 to 12VDC measurement range. Perhaps in the future it
could be expended for a wider range but the issue is getting very
accurate and stable (stable is the key operative issue) range divider
resistors. The design is based on the National Semiconductor app note
260, but modified with more modern components such as low charge
injection analog switches, faster comparators, LTZ1000 reference,
extremely linear integrator, low cost uprocessor, etc. My hope is
that the DVM will be able to measure to 8.5 digits over this limited
range. We shall see. It should be a fun project, regardless.
The big issue is the lack of noise rejection characteristics. A
substantial amount of averaging may be required so fast 8.5 digit
measurements might be out of the question. However, it should be a
low cost measurement technique for high accuracy measurements of
stable voltages. Obviously, the measurement accuracy would be largely
determined by the accuracy of the LTZ1000 reference.
Since the integrator is the key circuit for this design, I would
really like to see the integrator design in the HP-3458A, if anyone
could provide a copy of the circuit. Comments on the approach would
be apprecitated.
Hi Randy,
I have an ambition to do this too. But I think there are better circuit
arrangements that are not really any more complicated. Or at least it
seems that way right now, before I have tried to make them work!
Firstly you have to make a pretty good circuit before you can beat a
cheap single-chip sigma-delta ADC, as others have pointed out. That
application note predates these I think.
You could look at some of the patents on multislope converters (as used
in the HP3458A, and all the top meters as far as I know). They are
surprisingly readable. Also many of them have expired by now, although
that does not matter for educational use.
The "basic" HP patent is 4357600:
http://www.google.com/patents?id=vCUxAAAAEBAJ
I think this is the HP3458A.
There is a later HP one 4951053 which refines it with a greatly improved
switching technique http://www.google.com/patents?id=INImAAAAEBAJ. But
the 3458A does not use it AFAIK.
Then fluke get in the game with 5321403 and 5565869, describing a
switching sequence with imnproved elimination of charge injection
errors. (These have not expired).
There are lots of others if you search on "multislope" but I found these
the most interesting.
--
John Devereux
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
John,
Thanks for the info. I will read them tonight and assess wether I want to proceed with my original plan or revisit it. I have no particular preference, just what is easiest and cheapest but high resolution and accurate.
Randy
________________________________
From: John Devereux <john@devereux.me.uk>
To: volt-nuts@febo.com
Cc: Discussion of precise voltage measurement <volt-nuts@febo.com>
Sent: Sunday, November 13, 2011 11:42 PM
Subject: Re: [volt-nuts] Homebrew DVM
Randy Evans <randallgrayevans@yahoo.com> writes:
> I am in the process of designing a homebrew DVM that initially will
> only cover 0 to 12VDC measurement range. Perhaps in the future it
> could be expended for a wider range but the issue is getting very
> accurate and stable (stable is the key operative issue) range divider
> resistors. The design is based on the National Semiconductor app note
> 260, but modified with more modern components such as low charge
> injection analog switches, faster comparators, LTZ1000 reference,
> extremely linear integrator, low cost uprocessor, etc. My hope is
> that the DVM will be able to measure to 8.5 digits over this limited
> range. We shall see. It should be a fun project, regardless.
> The big issue is the lack of noise rejection characteristics. A
> substantial amount of averaging may be required so fast 8.5 digit
> measurements might be out of the question. However, it should be a
> low cost measurement technique for high accuracy measurements of
> stable voltages. Obviously, the measurement accuracy would be largely
> determined by the accuracy of the LTZ1000 reference.
> Since the integrator is the key circuit for this design, I would
> really like to see the integrator design in the HP-3458A, if anyone
> could provide a copy of the circuit. Comments on the approach would
> be apprecitated.
Hi Randy,
I have an ambition to do this too. But I think there are better circuit
arrangements that are not really any more complicated. Or at least it
seems that way right now, before I have tried to make them work!
Firstly you have to make a pretty good circuit before you can beat a
cheap single-chip sigma-delta ADC, as others have pointed out. That
application note predates these I think.
You could look at some of the patents on multislope converters (as used
in the HP3458A, and all the top meters as far as I know). They are
surprisingly readable. Also many of them have expired by now, although
that does not matter for educational use.
The "basic" HP patent is 4357600:
<http://www.google.com/patents?id=vCUxAAAAEBAJ>
I think this is the HP3458A.
There is a later HP one 4951053 which refines it with a greatly improved
switching technique <http://www.google.com/patents?id=INImAAAAEBAJ>. But
the 3458A does not use it AFAIK.
Then fluke get in the game with 5321403 and 5565869, describing a
switching sequence with imnproved elimination of charge injection
errors. (These have not expired).
There are lots of others if you search on "multislope" but I found these
the most interesting.
--
John Devereux
_______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
RS
Randy Scott
Mon, Nov 14, 2011 11:42 PM
Thanks for the info. I will read them tonight and assess whether
I want to proceed with my original plan or revisit it. I have
no particular preference, just what is easiest and cheapest
but high resolution and accurate.
The Wikipedia page on the integrating ADC also has some more
description on the basic dual-slope ADC as well as some of the
enhancements that have been made to it over the years. Some
of the material from the patents is mentioned as well.
http://en.wikipedia.org/wiki/Integrating_ADC
Many of the designs end up being quite simple, but the devil
is in the details, obviously.
What I find most interesting about the design is that, in many
cases, components with absolute precision are not required.
Stability and low TC are what is needed and sometimes the
ratio of certain critical values are important. Everything
else can be calibrated out at the end (within reason, of course).
Randy.
> Thanks for the info. I will read them tonight and assess whether
> I want to proceed with my original plan or revisit it. I have
> no particular preference, just what is easiest and cheapest
> but high resolution and accurate.
The Wikipedia page on the integrating ADC also has some more
description on the basic dual-slope ADC as well as some of the
enhancements that have been made to it over the years. Some
of the material from the patents is mentioned as well.
http://en.wikipedia.org/wiki/Integrating_ADC
Many of the designs end up being quite simple, but the devil
is in the details, obviously.
What I find most interesting about the design is that, in many
cases, components with absolute precision are not required.
Stability and low TC are what is needed and sometimes the
ratio of certain critical values are important. Everything
else can be calibrated out at the end (within reason, of course).
Randy.
FS
Fred Schneider
Mon, Nov 14, 2011 11:56 PM
I did some dual slope ADC experiments. Just on a breadboard and using an Arduino for the math and display. No fancy stuff. The VRef was just a resistive divider from the 5 V. It was just an experiment. Used a 4066, LF356 and a LM339 but the thing worked wonderfull. Got it stable up to 1 mV in a 1 V range. Not shocking but with a nice Vref, better switches, a real good capacitor it can be made better. While figuring out how that worked I misunderstood how it should work and by accident re-invented a ( self nulling) VF converter, using my counter as display. That was fun too.
Now I have a 26 bit ( 24 bit usable) ADC waiting to be used interfaced to the arduino, I am making/experimenting with some Vrefs and have parts for a compound chopper based on a LT1052. But the realisation will be up in the future, first gathering more knowledge about percision and micro volt techniques.
Fred PA4TIM
Op 15 nov. 2011 om 00:10 heeft Randy Evans randallgrayevans@yahoo.com het volgende geschreven:
John,
Thanks for the info. I will read them tonight and assess wether I want to proceed with my original plan or revisit it. I have no particular preference, just what is easiest and cheapest but high resolution and accurate.
Randy
From: John Devereux john@devereux.me.uk
To: volt-nuts@febo.com
Cc: Discussion of precise voltage measurement volt-nuts@febo.com
Sent: Sunday, November 13, 2011 11:42 PM
Subject: Re: [volt-nuts] Homebrew DVM
Randy Evans randallgrayevans@yahoo.com writes:
I am in the process of designing a homebrew DVM that initially will
only cover 0 to 12VDC measurement range. Perhaps in the future it
could be expended for a wider range but the issue is getting very
accurate and stable (stable is the key operative issue) range divider
resistors. The design is based on the National Semiconductor app note
260, but modified with more modern components such as low charge
injection analog switches, faster comparators, LTZ1000 reference,
extremely linear integrator, low cost uprocessor, etc. My hope is
that the DVM will be able to measure to 8.5 digits over this limited
range. We shall see. It should be a fun project, regardless.
The big issue is the lack of noise rejection characteristics. A
substantial amount of averaging may be required so fast 8.5 digit
measurements might be out of the question. However, it should be a
low cost measurement technique for high accuracy measurements of
stable voltages. Obviously, the measurement accuracy would be largely
determined by the accuracy of the LTZ1000 reference.
Since the integrator is the key circuit for this design, I would
really like to see the integrator design in the HP-3458A, if anyone
could provide a copy of the circuit. Comments on the approach would
be apprecitated.
Hi Randy,
I have an ambition to do this too. But I think there are better circuit
arrangements that are not really any more complicated. Or at least it
seems that way right now, before I have tried to make them work!
Firstly you have to make a pretty good circuit before you can beat a
cheap single-chip sigma-delta ADC, as others have pointed out. That
application note predates these I think.
You could look at some of the patents on multislope converters (as used
in the HP3458A, and all the top meters as far as I know). They are
surprisingly readable. Also many of them have expired by now, although
that does not matter for educational use.
The "basic" HP patent is 4357600:
http://www.google.com/patents?id=vCUxAAAAEBAJ
I think this is the HP3458A.
There is a later HP one 4951053 which refines it with a greatly improved
switching technique http://www.google.com/patents?id=INImAAAAEBAJ. But
the 3458A does not use it AFAIK.
Then fluke get in the game with 5321403 and 5565869, describing a
switching sequence with imnproved elimination of charge injection
errors. (These have not expired).
There are lots of others if you search on "multislope" but I found these
the most interesting.
--
John Devereux
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
I did some dual slope ADC experiments. Just on a breadboard and using an Arduino for the math and display. No fancy stuff. The VRef was just a resistive divider from the 5 V. It was just an experiment. Used a 4066, LF356 and a LM339 but the thing worked wonderfull. Got it stable up to 1 mV in a 1 V range. Not shocking but with a nice Vref, better switches, a real good capacitor it can be made better. While figuring out how that worked I misunderstood how it should work and by accident re-invented a ( self nulling) VF converter, using my counter as display. That was fun too.
Now I have a 26 bit ( 24 bit usable) ADC waiting to be used interfaced to the arduino, I am making/experimenting with some Vrefs and have parts for a compound chopper based on a LT1052. But the realisation will be up in the future, first gathering more knowledge about percision and micro volt techniques.
Fred PA4TIM
Op 15 nov. 2011 om 00:10 heeft Randy Evans <randallgrayevans@yahoo.com> het volgende geschreven:
> John,
>
> Thanks for the info. I will read them tonight and assess wether I want to proceed with my original plan or revisit it. I have no particular preference, just what is easiest and cheapest but high resolution and accurate.
>
> Randy
>
>
> ________________________________
> From: John Devereux <john@devereux.me.uk>
> To: volt-nuts@febo.com
> Cc: Discussion of precise voltage measurement <volt-nuts@febo.com>
> Sent: Sunday, November 13, 2011 11:42 PM
> Subject: Re: [volt-nuts] Homebrew DVM
>
> Randy Evans <randallgrayevans@yahoo.com> writes:
>
>> I am in the process of designing a homebrew DVM that initially will
>> only cover 0 to 12VDC measurement range. Perhaps in the future it
>> could be expended for a wider range but the issue is getting very
>> accurate and stable (stable is the key operative issue) range divider
>> resistors. The design is based on the National Semiconductor app note
>> 260, but modified with more modern components such as low charge
>> injection analog switches, faster comparators, LTZ1000 reference,
>> extremely linear integrator, low cost uprocessor, etc. My hope is
>> that the DVM will be able to measure to 8.5 digits over this limited
>> range. We shall see. It should be a fun project, regardless.
>
>> The big issue is the lack of noise rejection characteristics. A
>> substantial amount of averaging may be required so fast 8.5 digit
>> measurements might be out of the question. However, it should be a
>> low cost measurement technique for high accuracy measurements of
>> stable voltages. Obviously, the measurement accuracy would be largely
>> determined by the accuracy of the LTZ1000 reference.
>
>> Since the integrator is the key circuit for this design, I would
>> really like to see the integrator design in the HP-3458A, if anyone
>> could provide a copy of the circuit. Comments on the approach would
>> be apprecitated.
>
> Hi Randy,
>
> I have an ambition to do this too. But I think there are better circuit
> arrangements that are not really any more complicated. Or at least it
> seems that way right now, before I have tried to make them work!
>
> Firstly you have to make a pretty good circuit before you can beat a
> cheap single-chip sigma-delta ADC, as others have pointed out. That
> application note predates these I think.
>
> You could look at some of the patents on multislope converters (as used
> in the HP3458A, and all the top meters as far as I know). They are
> surprisingly readable. Also many of them have expired by now, although
> that does not matter for educational use.
>
> The "basic" HP patent is 4357600:
> <http://www.google.com/patents?id=vCUxAAAAEBAJ>
> I think this is the HP3458A.
>
> There is a later HP one 4951053 which refines it with a greatly improved
> switching technique <http://www.google.com/patents?id=INImAAAAEBAJ>. But
> the 3458A does not use it AFAIK.
>
> Then fluke get in the game with 5321403 and 5565869, describing a
> switching sequence with imnproved elimination of charge injection
> errors. (These have not expired).
>
> There are lots of others if you search on "multislope" but I found these
> the most interesting.
>
> --
>
> John Devereux
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
PK
Poul-Henning Kamp
Tue, Nov 15, 2011 12:17 AM
What I find most interesting about the design is that, in many
cases, components with absolute precision are not required.
Well, that's the fundamental trick of the integrating A/D: You
move the precision requirement to your timebase, and we are
much better at those, than at precision resistors/capacitors etc.
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
In message <1321314126.66316.YahooMailNeo@web161704.mail.bf1.yahoo.com>, Randy
Scott writes:
>What I find most interesting about the design is that, in many
>cases, components with absolute precision are not required.
Well, that's the fundamental trick of the integrating A/D: You
move the precision requirement to your timebase, and we are
much better at those, than at precision resistors/capacitors etc.
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
AJ
Andreas Jahn
Tue, Nov 15, 2011 5:38 AM
Got it stable up to 1 mV in a 1 V range.
Did you check only noise and temperature or also linearity?
My experience is that with a simple dual slope integrator the capacitor
soaking creates big linearity errors.
I built my first dual slope with a 2,49V LM336 reference a 4051 and a TLC271
and a simple 1uF mylar capacitor.
Due to single supply I used the 2,5V of the reference as vitual ground. So
the input range of 0..5V was
divided into two ranges from 0..2.5 (negative with respect to virtual
ground) and from 2.5 to 5V. (positive).
With a simple mylar capacitor the non-linear gap between negative and
positive range was around 1 mV.
So linearity is only around 12 bits with this design. To reach this
linearity I had to do some software
trick to deal with the soakage. After each switching of the input
multiplexer to another channel I did
some waste A/D conversions with the new channel to let the capacitor
stabilize.
Years later I read the article of Bob Pease about capacitor soaking and
exchanged the mylar against
polypropylene. The non-linear gap is now around 100uV wich is around 15 bits
linearity.
At the moment I'm experimenting with the 24 bit LTC2400.
He has a unipolar input range of 0..5V (+ some mV overrange) with a 5V
reference.
The best results I get with a LT1027CCN8-5 reference. From the many
references
I tried up to now the LT1027 and MAX6250A in DIL-housing seem to be the best
when regarding hysteresis.
The effort to get around 10 ppm stability is enormous:
Galvanic isolated battery powered design.
Temperature of reference is measured and compensated by a NTC (3rd order
polynominal compensation)
Linearity of A/D-Converter is compensated. (see AN86 of LT)
The main issue is noise of the converter of around 20uVpp.
So to get less than 1 uV I have to integrate measurements over 10 minutes.
Reference drift is around 10 ppm over 1 year.
I am looking now for a converter with less noise.
So what ADC are You using?
With best regards
Andreas
----- Original Message -----
From: "Fred Schneider" pa4tim@gmail.com
To: "Randy Evans" randallgrayevans@yahoo.com; "Discussion of
precisevoltage measurement" volt-nuts@febo.com
Sent: Tuesday, November 15, 2011 12:56 AM
Subject: Re: [volt-nuts] Homebrew DVM
I did some dual slope ADC experiments. Just on a breadboard and using an
Arduino for the math and display. No fancy stuff. The VRef was just a
resistive divider from the 5 V. It was just an experiment. Used a 4066,
LF356 and a LM339 but the thing worked wonderfull. Got it stable up to 1 mV
in a 1 V range. Not shocking but with a nice Vref, better switches, a real
good capacitor it can be made better. While figuring out how that worked I
misunderstood how it should work and by accident re-invented a ( self
nulling) VF converter, using my counter as display. That was fun too.
Now I have a 26 bit ( 24 bit usable) ADC waiting to be used interfaced to
the arduino, I am making/experimenting with some Vrefs and have parts for
a compound chopper based on a LT1052. But the realisation will be up in
the future, first gathering more knowledge about percision and micro volt
techniques.
Fred PA4TIM
Op 15 nov. 2011 om 00:10 heeft Randy Evans randallgrayevans@yahoo.com
het volgende geschreven:
John,
Thanks for the info. I will read them tonight and assess wether I want
to proceed with my original plan or revisit it. I have no particular
preference, just what is easiest and cheapest but high resolution and
accurate.
Randy
From: John Devereux john@devereux.me.uk
To: volt-nuts@febo.com
Cc: Discussion of precise voltage measurement volt-nuts@febo.com
Sent: Sunday, November 13, 2011 11:42 PM
Subject: Re: [volt-nuts] Homebrew DVM
Randy Evans randallgrayevans@yahoo.com writes:
I am in the process of designing a homebrew DVM that initially will
only cover 0 to 12VDC measurement range. Perhaps in the future it
could be expended for a wider range but the issue is getting very
accurate and stable (stable is the key operative issue) range divider
resistors. The design is based on the National Semiconductor app note
260, but modified with more modern components such as low charge
injection analog switches, faster comparators, LTZ1000 reference,
extremely linear integrator, low cost uprocessor, etc. My hope is
that the DVM will be able to measure to 8.5 digits over this limited
range. We shall see. It should be a fun project, regardless.
The big issue is the lack of noise rejection characteristics. A
substantial amount of averaging may be required so fast 8.5 digit
measurements might be out of the question. However, it should be a
low cost measurement technique for high accuracy measurements of
stable voltages. Obviously, the measurement accuracy would be largely
determined by the accuracy of the LTZ1000 reference.
Since the integrator is the key circuit for this design, I would
really like to see the integrator design in the HP-3458A, if anyone
could provide a copy of the circuit. Comments on the approach would
be apprecitated.
Hi Randy,
I have an ambition to do this too. But I think there are better circuit
arrangements that are not really any more complicated. Or at least it
seems that way right now, before I have tried to make them work!
Firstly you have to make a pretty good circuit before you can beat a
cheap single-chip sigma-delta ADC, as others have pointed out. That
application note predates these I think.
You could look at some of the patents on multislope converters (as used
in the HP3458A, and all the top meters as far as I know). They are
surprisingly readable. Also many of them have expired by now, although
that does not matter for educational use.
The "basic" HP patent is 4357600:
http://www.google.com/patents?id=vCUxAAAAEBAJ
I think this is the HP3458A.
There is a later HP one 4951053 which refines it with a greatly improved
switching technique http://www.google.com/patents?id=INImAAAAEBAJ. But
the 3458A does not use it AFAIK.
Then fluke get in the game with 5321403 and 5565869, describing a
switching sequence with imnproved elimination of charge injection
errors. (These have not expired).
There are lots of others if you search on "multislope" but I found these
the most interesting.
--
John Devereux
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
Hello Fred,
> Got it stable up to 1 mV in a 1 V range.
Did you check only noise and temperature or also linearity?
My experience is that with a simple dual slope integrator the capacitor
soaking creates big linearity errors.
I built my first dual slope with a 2,49V LM336 reference a 4051 and a TLC271
and a simple 1uF mylar capacitor.
Due to single supply I used the 2,5V of the reference as vitual ground. So
the input range of 0..5V was
divided into two ranges from 0..2.5 (negative with respect to virtual
ground) and from 2.5 to 5V. (positive).
With a simple mylar capacitor the non-linear gap between negative and
positive range was around 1 mV.
So linearity is only around 12 bits with this design. To reach this
linearity I had to do some software
trick to deal with the soakage. After each switching of the input
multiplexer to another channel I did
some waste A/D conversions with the new channel to let the capacitor
stabilize.
Years later I read the article of Bob Pease about capacitor soaking and
exchanged the mylar against
polypropylene. The non-linear gap is now around 100uV wich is around 15 bits
linearity.
At the moment I'm experimenting with the 24 bit LTC2400.
He has a unipolar input range of 0..5V (+ some mV overrange) with a 5V
reference.
The best results I get with a LT1027CCN8-5 reference. From the many
references
I tried up to now the LT1027 and MAX6250A in DIL-housing seem to be the best
when regarding hysteresis.
The effort to get around 10 ppm stability is enormous:
Galvanic isolated battery powered design.
Temperature of reference is measured and compensated by a NTC (3rd order
polynominal compensation)
Linearity of A/D-Converter is compensated. (see AN86 of LT)
The main issue is noise of the converter of around 20uVpp.
So to get less than 1 uV I have to integrate measurements over 10 minutes.
Reference drift is around 10 ppm over 1 year.
I am looking now for a converter with less noise.
So what ADC are You using?
With best regards
Andreas
----- Original Message -----
From: "Fred Schneider" <pa4tim@gmail.com>
To: "Randy Evans" <randallgrayevans@yahoo.com>; "Discussion of
precisevoltage measurement" <volt-nuts@febo.com>
Sent: Tuesday, November 15, 2011 12:56 AM
Subject: Re: [volt-nuts] Homebrew DVM
>I did some dual slope ADC experiments. Just on a breadboard and using an
>Arduino for the math and display. No fancy stuff. The VRef was just a
>resistive divider from the 5 V. It was just an experiment. Used a 4066,
>LF356 and a LM339 but the thing worked wonderfull. Got it stable up to 1 mV
>in a 1 V range. Not shocking but with a nice Vref, better switches, a real
>good capacitor it can be made better. While figuring out how that worked I
>misunderstood how it should work and by accident re-invented a ( self
>nulling) VF converter, using my counter as display. That was fun too.
>
> Now I have a 26 bit ( 24 bit usable) ADC waiting to be used interfaced to
> the arduino, I am making/experimenting with some Vrefs and have parts for
> a compound chopper based on a LT1052. But the realisation will be up in
> the future, first gathering more knowledge about percision and micro volt
> techniques.
>
> Fred PA4TIM
>
> Op 15 nov. 2011 om 00:10 heeft Randy Evans <randallgrayevans@yahoo.com>
> het volgende geschreven:
>
>> John,
>>
>> Thanks for the info. I will read them tonight and assess wether I want
>> to proceed with my original plan or revisit it. I have no particular
>> preference, just what is easiest and cheapest but high resolution and
>> accurate.
>>
>> Randy
>>
>>
>> ________________________________
>> From: John Devereux <john@devereux.me.uk>
>> To: volt-nuts@febo.com
>> Cc: Discussion of precise voltage measurement <volt-nuts@febo.com>
>> Sent: Sunday, November 13, 2011 11:42 PM
>> Subject: Re: [volt-nuts] Homebrew DVM
>>
>> Randy Evans <randallgrayevans@yahoo.com> writes:
>>
>>> I am in the process of designing a homebrew DVM that initially will
>>> only cover 0 to 12VDC measurement range. Perhaps in the future it
>>> could be expended for a wider range but the issue is getting very
>>> accurate and stable (stable is the key operative issue) range divider
>>> resistors. The design is based on the National Semiconductor app note
>>> 260, but modified with more modern components such as low charge
>>> injection analog switches, faster comparators, LTZ1000 reference,
>>> extremely linear integrator, low cost uprocessor, etc. My hope is
>>> that the DVM will be able to measure to 8.5 digits over this limited
>>> range. We shall see. It should be a fun project, regardless.
>>
>>> The big issue is the lack of noise rejection characteristics. A
>>> substantial amount of averaging may be required so fast 8.5 digit
>>> measurements might be out of the question. However, it should be a
>>> low cost measurement technique for high accuracy measurements of
>>> stable voltages. Obviously, the measurement accuracy would be largely
>>> determined by the accuracy of the LTZ1000 reference.
>>
>>> Since the integrator is the key circuit for this design, I would
>>> really like to see the integrator design in the HP-3458A, if anyone
>>> could provide a copy of the circuit. Comments on the approach would
>>> be apprecitated.
>>
>> Hi Randy,
>>
>> I have an ambition to do this too. But I think there are better circuit
>> arrangements that are not really any more complicated. Or at least it
>> seems that way right now, before I have tried to make them work!
>>
>> Firstly you have to make a pretty good circuit before you can beat a
>> cheap single-chip sigma-delta ADC, as others have pointed out. That
>> application note predates these I think.
>>
>> You could look at some of the patents on multislope converters (as used
>> in the HP3458A, and all the top meters as far as I know). They are
>> surprisingly readable. Also many of them have expired by now, although
>> that does not matter for educational use.
>>
>> The "basic" HP patent is 4357600:
>> <http://www.google.com/patents?id=vCUxAAAAEBAJ>
>> I think this is the HP3458A.
>>
>> There is a later HP one 4951053 which refines it with a greatly improved
>> switching technique <http://www.google.com/patents?id=INImAAAAEBAJ>. But
>> the 3458A does not use it AFAIK.
>>
>> Then fluke get in the game with 5321403 and 5565869, describing a
>> switching sequence with imnproved elimination of charge injection
>> errors. (These have not expired).
>>
>> There are lots of others if you search on "multislope" but I found these
>> the most interesting.
>>
>> --
>>
>> John Devereux
>>
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
BG
Bruce Griffiths
Tue, Nov 15, 2011 5:54 AM
Got it stable up to 1 mV in a 1 V range.
Did you check only noise and temperature or also linearity?
My experience is that with a simple dual slope integrator the
capacitor soaking creates big linearity errors.
I built my first dual slope with a 2,49V LM336 reference a 4051 and a
TLC271 and a simple 1uF mylar capacitor.
Due to single supply I used the 2,5V of the reference as vitual
ground. So the input range of 0..5V was
divided into two ranges from 0..2.5 (negative with respect to virtual
ground) and from 2.5 to 5V. (positive).
With a simple mylar capacitor the non-linear gap between negative and
positive range was around 1 mV.
So linearity is only around 12 bits with this design. To reach this
linearity I had to do some software
trick to deal with the soakage. After each switching of the input
multiplexer to another channel I did
some waste A/D conversions with the new channel to let the capacitor
stabilize.
Years later I read the article of Bob Pease about capacitor soaking
and exchanged the mylar against
polypropylene. The non-linear gap is now around 100uV wich is around
15 bits linearity.
The HP3458A and the HP34401A ensure that the integrator output voltage
only represents a small fraction of the integral of the signal input.
This significantly reduces the effect of dielectric absorption in the
integrator feedback capacitor.
The integrator and control logic act somewhat like a sigma-delta
integrator during the runup phase.
The effect of dielectric absorption can be further reduced by suitable
choice of runup algorithm.
The on resistance modulation of analog switches in series with the
resistor connected between the signal and the integrator summing
junction can be a significant source of nonlinearity, however it is, in
principle, possible to correct for this.
Bruce
Andreas Jahn wrote:
> Hello Fred,
>
>> Got it stable up to 1 mV in a 1 V range.
>
> Did you check only noise and temperature or also linearity?
>
> My experience is that with a simple dual slope integrator the
> capacitor soaking creates big linearity errors.
> I built my first dual slope with a 2,49V LM336 reference a 4051 and a
> TLC271 and a simple 1uF mylar capacitor.
> Due to single supply I used the 2,5V of the reference as vitual
> ground. So the input range of 0..5V was
> divided into two ranges from 0..2.5 (negative with respect to virtual
> ground) and from 2.5 to 5V. (positive).
> With a simple mylar capacitor the non-linear gap between negative and
> positive range was around 1 mV.
>
> So linearity is only around 12 bits with this design. To reach this
> linearity I had to do some software
> trick to deal with the soakage. After each switching of the input
> multiplexer to another channel I did
> some waste A/D conversions with the new channel to let the capacitor
> stabilize.
>
> Years later I read the article of Bob Pease about capacitor soaking
> and exchanged the mylar against
> polypropylene. The non-linear gap is now around 100uV wich is around
> 15 bits linearity.
The HP3458A and the HP34401A ensure that the integrator output voltage
only represents a small fraction of the integral of the signal input.
This significantly reduces the effect of dielectric absorption in the
integrator feedback capacitor.
The integrator and control logic act somewhat like a sigma-delta
integrator during the runup phase.
The effect of dielectric absorption can be further reduced by suitable
choice of runup algorithm.
The on resistance modulation of analog switches in series with the
resistor connected between the signal and the integrator summing
junction can be a significant source of nonlinearity, however it is, in
principle, possible to correct for this.
Bruce
FS
Fred Schneider
Tue, Nov 15, 2011 7:46 AM
No, did not check all. Just hooked up a powersupply and looked what it did. It was at first absolute not lineair. Before it was finished my ardiuno went up in smoke because of the breadboard ( +5 V jumped out and I did not think and just pushed it back in the +15V rail) but It was not very lineair until I made a sort of autozero and used a correction factor. After that it was sort of lineair, I had made it so that it did a measurement of the test-voltage then It switched the input to ground and made a second measurement. That value I used to correct the other.
It was just for fun, to learn about the principle of ADCs. But it infected me with the volt nut virus.
It was not intended to become a real meter.
http://www.pa4tim.nl/?p=2482 there you see the schematic and some pictures ( the top one is from the circuit that, i was told, turned out to be a VF converter)
Fred PA4TIM
Op 15 nov. 2011 om 06:38 heeft "Andreas Jahn" Andreas_-_Jahn@t-online.de het volgende geschreven:
Got it stable up to 1 mV in a 1 V range.
Did you check only noise and temperature or also linearity?
My experience is that with a simple dual slope integrator the capacitor soaking creates big linearity errors.
I built my first dual slope with a 2,49V LM336 reference a 4051 and a TLC271 and a simple 1uF mylar capacitor.
Due to single supply I used the 2,5V of the reference as vitual ground. So the input range of 0..5V was
divided into two ranges from 0..2.5 (negative with respect to virtual ground) and from 2.5 to 5V. (positive).
With a simple mylar capacitor the non-linear gap between negative and positive range was around 1 mV.
So linearity is only around 12 bits with this design. To reach this linearity I had to do some software
trick to deal with the soakage. After each switching of the input multiplexer to another channel I did
some waste A/D conversions with the new channel to let the capacitor stabilize.
Years later I read the article of Bob Pease about capacitor soaking and exchanged the mylar against
polypropylene. The non-linear gap is now around 100uV wich is around 15 bits linearity.
At the moment I'm experimenting with the 24 bit LTC2400.
He has a unipolar input range of 0..5V (+ some mV overrange) with a 5V reference.
The best results I get with a LT1027CCN8-5 reference. From the many references
I tried up to now the LT1027 and MAX6250A in DIL-housing seem to be the best
when regarding hysteresis.
The effort to get around 10 ppm stability is enormous:
Galvanic isolated battery powered design.
Temperature of reference is measured and compensated by a NTC (3rd order polynominal compensation)
Linearity of A/D-Converter is compensated. (see AN86 of LT)
The main issue is noise of the converter of around 20uVpp.
So to get less than 1 uV I have to integrate measurements over 10 minutes.
Reference drift is around 10 ppm over 1 year.
I am looking now for a converter with less noise.
So what ADC are You using?
With best regards
Andreas
----- Original Message ----- From: "Fred Schneider" pa4tim@gmail.com
To: "Randy Evans" randallgrayevans@yahoo.com; "Discussion of precisevoltage measurement" volt-nuts@febo.com
Sent: Tuesday, November 15, 2011 12:56 AM
Subject: Re: [volt-nuts] Homebrew DVM
I did some dual slope ADC experiments. Just on a breadboard and using an Arduino for the math and display. No fancy stuff. The VRef was just a resistive divider from the 5 V. It was just an experiment. Used a 4066, LF356 and a LM339 but the thing worked wonderfull. Got it stable up to 1 mV in a 1 V range. Not shocking but with a nice Vref, better switches, a real good capacitor it can be made better. While figuring out how that worked I misunderstood how it should work and by accident re-invented a ( self nulling) VF converter, using my counter as display. That was fun too.
Now I have a 26 bit ( 24 bit usable) ADC waiting to be used interfaced to the arduino, I am making/experimenting with some Vrefs and have parts for a compound chopper based on a LT1052. But the realisation will be up in the future, first gathering more knowledge about percision and micro volt techniques.
Fred PA4TIM
Op 15 nov. 2011 om 00:10 heeft Randy Evans randallgrayevans@yahoo.com het volgende geschreven:
John,
Thanks for the info. I will read them tonight and assess wether I want to proceed with my original plan or revisit it. I have no particular preference, just what is easiest and cheapest but high resolution and accurate.
Randy
From: John Devereux john@devereux.me.uk
To: volt-nuts@febo.com
Cc: Discussion of precise voltage measurement volt-nuts@febo.com
Sent: Sunday, November 13, 2011 11:42 PM
Subject: Re: [volt-nuts] Homebrew DVM
Randy Evans randallgrayevans@yahoo.com writes:
I am in the process of designing a homebrew DVM that initially will
only cover 0 to 12VDC measurement range. Perhaps in the future it
could be expended for a wider range but the issue is getting very
accurate and stable (stable is the key operative issue) range divider
resistors. The design is based on the National Semiconductor app note
260, but modified with more modern components such as low charge
injection analog switches, faster comparators, LTZ1000 reference,
extremely linear integrator, low cost uprocessor, etc. My hope is
that the DVM will be able to measure to 8.5 digits over this limited
range. We shall see. It should be a fun project, regardless.
The big issue is the lack of noise rejection characteristics. A
substantial amount of averaging may be required so fast 8.5 digit
measurements might be out of the question. However, it should be a
low cost measurement technique for high accuracy measurements of
stable voltages. Obviously, the measurement accuracy would be largely
determined by the accuracy of the LTZ1000 reference.
Since the integrator is the key circuit for this design, I would
really like to see the integrator design in the HP-3458A, if anyone
could provide a copy of the circuit. Comments on the approach would
be apprecitated.
Hi Randy,
I have an ambition to do this too. But I think there are better circuit
arrangements that are not really any more complicated. Or at least it
seems that way right now, before I have tried to make them work!
Firstly you have to make a pretty good circuit before you can beat a
cheap single-chip sigma-delta ADC, as others have pointed out. That
application note predates these I think.
You could look at some of the patents on multislope converters (as used
in the HP3458A, and all the top meters as far as I know). They are
surprisingly readable. Also many of them have expired by now, although
that does not matter for educational use.
The "basic" HP patent is 4357600:
http://www.google.com/patents?id=vCUxAAAAEBAJ
I think this is the HP3458A.
There is a later HP one 4951053 which refines it with a greatly improved
switching technique http://www.google.com/patents?id=INImAAAAEBAJ. But
the 3458A does not use it AFAIK.
Then fluke get in the game with 5321403 and 5565869, describing a
switching sequence with imnproved elimination of charge injection
errors. (These have not expired).
There are lots of others if you search on "multislope" but I found these
the most interesting.
--
John Devereux
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
No, did not check all. Just hooked up a powersupply and looked what it did. It was at first absolute not lineair. Before it was finished my ardiuno went up in smoke because of the breadboard ( +5 V jumped out and I did not think and just pushed it back in the +15V rail) but It was not very lineair until I made a sort of autozero and used a correction factor. After that it was sort of lineair, I had made it so that it did a measurement of the test-voltage then It switched the input to ground and made a second measurement. That value I used to correct the other.
It was just for fun, to learn about the principle of ADCs. But it infected me with the volt nut virus.
It was not intended to become a real meter.
http://www.pa4tim.nl/?p=2482 there you see the schematic and some pictures ( the top one is from the circuit that, i was told, turned out to be a VF converter)
Fred PA4TIM
Op 15 nov. 2011 om 06:38 heeft "Andreas Jahn" <Andreas_-_Jahn@t-online.de> het volgende geschreven:
> Hello Fred,
>
>> Got it stable up to 1 mV in a 1 V range.
>
> Did you check only noise and temperature or also linearity?
>
> My experience is that with a simple dual slope integrator the capacitor soaking creates big linearity errors.
> I built my first dual slope with a 2,49V LM336 reference a 4051 and a TLC271 and a simple 1uF mylar capacitor.
> Due to single supply I used the 2,5V of the reference as vitual ground. So the input range of 0..5V was
> divided into two ranges from 0..2.5 (negative with respect to virtual ground) and from 2.5 to 5V. (positive).
> With a simple mylar capacitor the non-linear gap between negative and positive range was around 1 mV.
>
> So linearity is only around 12 bits with this design. To reach this linearity I had to do some software
> trick to deal with the soakage. After each switching of the input multiplexer to another channel I did
> some waste A/D conversions with the new channel to let the capacitor stabilize.
>
> Years later I read the article of Bob Pease about capacitor soaking and exchanged the mylar against
> polypropylene. The non-linear gap is now around 100uV wich is around 15 bits linearity.
>
> At the moment I'm experimenting with the 24 bit LTC2400.
> He has a unipolar input range of 0..5V (+ some mV overrange) with a 5V reference.
> The best results I get with a LT1027CCN8-5 reference. From the many references
> I tried up to now the LT1027 and MAX6250A in DIL-housing seem to be the best
> when regarding hysteresis.
>
> The effort to get around 10 ppm stability is enormous:
> Galvanic isolated battery powered design.
> Temperature of reference is measured and compensated by a NTC (3rd order polynominal compensation)
> Linearity of A/D-Converter is compensated. (see AN86 of LT)
>
> The main issue is noise of the converter of around 20uVpp.
> So to get less than 1 uV I have to integrate measurements over 10 minutes.
> Reference drift is around 10 ppm over 1 year.
>
> I am looking now for a converter with less noise.
> So what ADC are You using?
>
> With best regards
>
> Andreas
>
>
>
> ----- Original Message ----- From: "Fred Schneider" <pa4tim@gmail.com>
> To: "Randy Evans" <randallgrayevans@yahoo.com>; "Discussion of precisevoltage measurement" <volt-nuts@febo.com>
> Sent: Tuesday, November 15, 2011 12:56 AM
> Subject: Re: [volt-nuts] Homebrew DVM
>
>
>> I did some dual slope ADC experiments. Just on a breadboard and using an Arduino for the math and display. No fancy stuff. The VRef was just a resistive divider from the 5 V. It was just an experiment. Used a 4066, LF356 and a LM339 but the thing worked wonderfull. Got it stable up to 1 mV in a 1 V range. Not shocking but with a nice Vref, better switches, a real good capacitor it can be made better. While figuring out how that worked I misunderstood how it should work and by accident re-invented a ( self nulling) VF converter, using my counter as display. That was fun too.
>>
>> Now I have a 26 bit ( 24 bit usable) ADC waiting to be used interfaced to the arduino, I am making/experimenting with some Vrefs and have parts for a compound chopper based on a LT1052. But the realisation will be up in the future, first gathering more knowledge about percision and micro volt techniques.
>>
>> Fred PA4TIM
>>
>> Op 15 nov. 2011 om 00:10 heeft Randy Evans <randallgrayevans@yahoo.com> het volgende geschreven:
>>
>>> John,
>>>
>>> Thanks for the info. I will read them tonight and assess wether I want to proceed with my original plan or revisit it. I have no particular preference, just what is easiest and cheapest but high resolution and accurate.
>>>
>>> Randy
>>>
>>>
>>> ________________________________
>>> From: John Devereux <john@devereux.me.uk>
>>> To: volt-nuts@febo.com
>>> Cc: Discussion of precise voltage measurement <volt-nuts@febo.com>
>>> Sent: Sunday, November 13, 2011 11:42 PM
>>> Subject: Re: [volt-nuts] Homebrew DVM
>>>
>>> Randy Evans <randallgrayevans@yahoo.com> writes:
>>>
>>>> I am in the process of designing a homebrew DVM that initially will
>>>> only cover 0 to 12VDC measurement range. Perhaps in the future it
>>>> could be expended for a wider range but the issue is getting very
>>>> accurate and stable (stable is the key operative issue) range divider
>>>> resistors. The design is based on the National Semiconductor app note
>>>> 260, but modified with more modern components such as low charge
>>>> injection analog switches, faster comparators, LTZ1000 reference,
>>>> extremely linear integrator, low cost uprocessor, etc. My hope is
>>>> that the DVM will be able to measure to 8.5 digits over this limited
>>>> range. We shall see. It should be a fun project, regardless.
>>>
>>>> The big issue is the lack of noise rejection characteristics. A
>>>> substantial amount of averaging may be required so fast 8.5 digit
>>>> measurements might be out of the question. However, it should be a
>>>> low cost measurement technique for high accuracy measurements of
>>>> stable voltages. Obviously, the measurement accuracy would be largely
>>>> determined by the accuracy of the LTZ1000 reference.
>>>
>>>> Since the integrator is the key circuit for this design, I would
>>>> really like to see the integrator design in the HP-3458A, if anyone
>>>> could provide a copy of the circuit. Comments on the approach would
>>>> be apprecitated.
>>>
>>> Hi Randy,
>>>
>>> I have an ambition to do this too. But I think there are better circuit
>>> arrangements that are not really any more complicated. Or at least it
>>> seems that way right now, before I have tried to make them work!
>>>
>>> Firstly you have to make a pretty good circuit before you can beat a
>>> cheap single-chip sigma-delta ADC, as others have pointed out. That
>>> application note predates these I think.
>>>
>>> You could look at some of the patents on multislope converters (as used
>>> in the HP3458A, and all the top meters as far as I know). They are
>>> surprisingly readable. Also many of them have expired by now, although
>>> that does not matter for educational use.
>>>
>>> The "basic" HP patent is 4357600:
>>> <http://www.google.com/patents?id=vCUxAAAAEBAJ>
>>> I think this is the HP3458A.
>>>
>>> There is a later HP one 4951053 which refines it with a greatly improved
>>> switching technique <http://www.google.com/patents?id=INImAAAAEBAJ>. But
>>> the 3458A does not use it AFAIK.
>>>
>>> Then fluke get in the game with 5321403 and 5565869, describing a
>>> switching sequence with imnproved elimination of charge injection
>>> errors. (These have not expired).
>>>
>>> There are lots of others if you search on "multislope" but I found these
>>> the most interesting.
>>>
>>> --
>>>
>>> John Devereux
>>>
>>> _______________________________________________
>>> volt-nuts mailing list -- volt-nuts@febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>> and follow the instructions there.
>>> _______________________________________________
>>> volt-nuts mailing list -- volt-nuts@febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>>> and follow the instructions there.
>>
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
>> and follow the instructions there.
>
>
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.