Hello Time-Nuts,
I recently acquired an HP 5065A rubidium oscillator (with 10811 10MHz
OCXO). I think I pretty much have it running now and have been letting it
cook for the last couple of weeks. I offset the C-field + and - and
measured the frequency to calculate the C-field sensitivity. My unit came
out to 1.96x10E-12 per dial unit, which agrees with the manual stated
2x10E-12. So, calculated the on frequency C-field value and dialed it in.
Attached is a Lady Heather plot of the frequency over the last 3 days. The
purple line is the 1pps plot with the vertical scale being 20ns per
division. So, the unit is off about 125ns over the last 72 hours (running
about 4.92x10E-13 slow). So my C-field setting is off about 1/4 of a
division, but I think I'm going to leave well enough alone.
The yellow line is the NTBW50AA temperature sensor, and you can clearly see
when the furnace cycles. I was away for the weekend, and you can also
clearly see when I came home this evening and turned up the heat. At the
very end of the plot is the spike when I turned on the lights in the shop.
I love using a Thunderbolt/NTBW50AA for making frequency measurements this
way. I remove the OCXO, and insert the 10MHz from the DUT. Then disable
disciplining so the DAC voltage doesn't try to chase the open loop
oscillator. Of course the short-term performance looks worse than it
actually is because of measuring against GPS, but the long-term
measurements are very good.
I want to log this unit at regular intervals to see what the aging looks
like. Also need to do some measurements against the cesium to see what the
short-term performance might be. But, I think this oscillator will be a
good reference in many cases in lieu of using the cesium.
Regards,
Skip Withrow
On 11/08/2016 12:16 AM, Skip Withrow wrote:
Hello Time-Nuts,
I recently acquired an HP 5065A rubidium oscillator (with 10811 10MHz
OCXO). I think I pretty much have it running now and have been letting it
cook for the last couple of weeks. I offset the C-field + and - and
measured the frequency to calculate the C-field sensitivity. My unit came
out to 1.96x10E-12 per dial unit, which agrees with the manual stated
2x10E-12. So, calculated the on frequency C-field value and dialed it in.
Attached is a Lady Heather plot of the frequency over the last 3 days. The
purple line is the 1pps plot with the vertical scale being 20ns per
division. So, the unit is off about 125ns over the last 72 hours (running
about 4.92x10E-13 slow). So my C-field setting is off about 1/4 of a
division, but I think I'm going to leave well enough alone.
The yellow line is the NTBW50AA temperature sensor, and you can clearly see
when the furnace cycles. I was away for the weekend, and you can also
clearly see when I came home this evening and turned up the heat. At the
very end of the plot is the spike when I turned on the lights in the shop.
I love using a Thunderbolt/NTBW50AA for making frequency measurements this
way. I remove the OCXO, and insert the 10MHz from the DUT. Then disable
disciplining so the DAC voltage doesn't try to chase the open loop
oscillator. Of course the short-term performance looks worse than it
actually is because of measuring against GPS, but the long-term
measurements are very good.
I want to log this unit at regular intervals to see what the aging looks
like. Also need to do some measurements against the cesium to see what the
short-term performance might be. But, I think this oscillator will be a
good reference in many cases in lieu of using the cesium.
Regards,
Skip Withrow
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
I hope you will forgive me for reducing the SNR of this forum a bit, but
is that a pro-Trump message embedded in that Lady Heather plot?
Furthermore, it might also be promoting electoral fraud, though
``...vote often!'' is ambiguous (maybe that's plausible deniability?)
-Ruslan
Skip
Wow a poor mans test setup. Kind of my speed and no HPIB or big counters
etc. Like the re-use of Lady Heather. Clever.
OK now to go find the soldering Iron and hack my TBolt... Maybe not today.
So if you have a 5065 with 5 MHz out it needs to be doubled?
With respect to the cfield. Yes its pretty sensitive down in the sub minor
tick area.
When I had the 5071 for testing I could never get it quite right. But
nothing was actually wrong or noisy it was at the limit of reality.
Regards
Paul.
WB8TSL
On Tue, Nov 8, 2016 at 8:12 AM, Ruslan Nabioullin rnabioullin@gmail.com
wrote:
On 11/08/2016 12:16 AM, Skip Withrow wrote:
Hello Time-Nuts,
I recently acquired an HP 5065A rubidium oscillator (with 10811 10MHz
OCXO). I think I pretty much have it running now and have been letting it
cook for the last couple of weeks. I offset the C-field + and - and
measured the frequency to calculate the C-field sensitivity. My unit came
out to 1.96x10E-12 per dial unit, which agrees with the manual stated
2x10E-12. So, calculated the on frequency C-field value and dialed it in.
Attached is a Lady Heather plot of the frequency over the last 3 days.
The
purple line is the 1pps plot with the vertical scale being 20ns per
division. So, the unit is off about 125ns over the last 72 hours (running
about 4.92x10E-13 slow). So my C-field setting is off about 1/4 of a
division, but I think I'm going to leave well enough alone.
The yellow line is the NTBW50AA temperature sensor, and you can clearly
see
when the furnace cycles. I was away for the weekend, and you can also
clearly see when I came home this evening and turned up the heat. At the
very end of the plot is the spike when I turned on the lights in the shop.
I love using a Thunderbolt/NTBW50AA for making frequency measurements this
way. I remove the OCXO, and insert the 10MHz from the DUT. Then disable
disciplining so the DAC voltage doesn't try to chase the open loop
oscillator. Of course the short-term performance looks worse than it
actually is because of measuring against GPS, but the long-term
measurements are very good.
I want to log this unit at regular intervals to see what the aging looks
like. Also need to do some measurements against the cesium to see what
the
short-term performance might be. But, I think this oscillator will be a
good reference in many cases in lieu of using the cesium.
Regards,
Skip Withrow
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.
I hope you will forgive me for reducing the SNR of this forum a bit, but
is that a pro-Trump message embedded in that Lady Heather plot?
Furthermore, it might also be promoting electoral fraud, though ``...vote
often!'' is ambiguous (maybe that's plausible deniability?)
-Ruslan
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.