Hello All Again,
I’ve got my 5370B going now and in the process made a “discovery” which I thought might be worth sharing:
The A3 input board is a through-hole PCB with a few SMD capacitors and resistors on the reverse of the “switch area”. It turns out that on my 5370 (2410A00777) these components are not soldered, but fitted with conductive adhesive. I first thought it was solder with a black coating but under a microscope it is clear that it is not solder at all. Most probably it is a mixture of epoxy and silver particles, or a similar compound. So no going over joints with a fine iron…
Inspecting all this carefully under the microscope I discovered that the “joints” on 2 resistors (R23 and R56) had cracked. As you know this board gets heat from the hybrid amplifier IC’s and due to the way the board is mounted to the front panel I guess it sees thermal stressing when the instrument warms up and cools down. While this obviously lasts a long time it looks that on my unit the adhesive has eventually cracked in places. (Vigorous switch activation and pressing / pulling on the switch handles also won’t be helpful in this respect…) One of the resistors just fell off at the slightest touch with fine tweezers.
Anyhow, after removing the offending components, cleaning the pads of the adhesive, and soldering replacements in place, we have a perfect 99.9x ns with the 10 MHz on the commoned inputs. Happy days!
So if any of your 5370’s have the kind of intermittent fault I described (and one or two other people seem to have reported) or instability that seems to originate from the A3 board – check the joints around the SMD’s.
I wonder why / how it ended up having the adhesive instead of solder – were earlier / later instruments the same, or was this a build change introduced at a certain period?
Hope the above is of help to somebody else,
Thomas.
Message: 1
Date: Sat, 15 Jul 2017 21:54:47 +0100
From: Thomas Allgeier th.allgeier@gmail.com
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Subject: Re: [time-nuts] 5370B Question / help needed
Message-ID:
CAOT1cQLOjZ0=Xs_+nQ+MzAODboZG5ar0f8G_PydxJ=Oj6grB8g@mail.gmail.com
Content-Type: text/plain; charset="UTF-8"
Hello Folks,
I'm coming back to this after a few days of prodding etc. and there has
been some progress. The fault has been tracked down to the A3 input board
and it points at a crack in the board or a joint. Here is the sketch:
After a good clean of the switches and re-seating all boards etc. the 5370
sprung back into life with almost-in-spec performance of 102.xx ns with
it's own 10 MHz. Trouble is it didn't last - after a while and triggered by
sliding some switches around it went back to the 14 ns. Fiddling with the
BNC's may have the same effect, i.e. there is a mechanical element to it
all.
Turns out that running it with the front panel removed I can make it go
from 14 to 102 ns by slightly bending the A3 board, certainly while it is
cold. After a while this trick doesn't work anymore, my suspicion is
whatever crack/gap is causing the trouble has expanded too far to close it
tight. After cool-down we're back to square one.
I notice there are a few SMD components on that board, right in the middle
where it would bend most - basically on the reverse side of the switches.
Capacitors and resistors I expect. I wonder if it is worth going over the
solder joints of all these carefully.
What has stopped me so far are 2 questions:
Am I the only one with this observation or has anybody come across this
before?
Secondly the solder on these SMD's is coated with a black substance.
Clearly this could be removed somehow but it is probably there for a
reason. Has anybody re-touched joints like these before?
On an instrument like this it is very much a case of "proceed with care"
and I'd hate to do more damage by rushing in.
On a related subject: last year there was a discussion over redesigning the
input board(s) for the 5370 / 5345. Did this get off the ground? If my A3
packs up completely...
Thanks in advance for any suggestions,
Thomas.
This is very interesting, I have a 5370B that has the same behavior as
yours. I will take the input board out again and give it a good
inspection. BTW, what did you use to clean the adhesive from the pads ?
Thanks
Gary
On 7/18/2017 2:41 AM, Thomas Allgeier wrote:
Hello All Again,
I’ve got my 5370B going now and in the process made a “discovery” which I thought might be worth sharing:
The A3 input board is a through-hole PCB with a few SMD capacitors and resistors on the reverse of the “switch area”. It turns out that on my 5370 (2410A00777) these components are not soldered, but fitted with conductive adhesive. I first thought it was solder with a black coating but under a microscope it is clear that it is not solder at all. Most probably it is a mixture of epoxy and silver particles, or a similar compound. So no going over joints with a fine iron…
Inspecting all this carefully under the microscope I discovered that the “joints” on 2 resistors (R23 and R56) had cracked. As you know this board gets heat from the hybrid amplifier IC’s and due to the way the board is mounted to the front panel I guess it sees thermal stressing when the instrument warms up and cools down. While this obviously lasts a long time it looks that on my unit the adhesive has eventually cracked in places. (Vigorous switch activation and pressing / pulling on the switch handles also won’t be helpful in this respect…) One of the resistors just fell off at the slightest touch with fine tweezers.
Anyhow, after removing the offending components, cleaning the pads of the adhesive, and soldering replacements in place, we have a perfect 99.9x ns with the 10 MHz on the commoned inputs. Happy days!
So if any of your 5370’s have the kind of intermittent fault I described (and one or two other people seem to have reported) or instability that seems to originate from the A3 board – check the joints around the SMD’s.
I wonder why / how it ended up having the adhesive instead of solder – were earlier / later instruments the same, or was this a build change introduced at a certain period?
Hope the above is of help to somebody else,
Thomas.
Message: 1
Date: Sat, 15 Jul 2017 21:54:47 +0100
From: Thomas Allgeier th.allgeier@gmail.com
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Subject: Re: [time-nuts] 5370B Question / help needed
Message-ID:
CAOT1cQLOjZ0=Xs_+nQ+MzAODboZG5ar0f8G_PydxJ=Oj6grB8g@mail.gmail.com
Content-Type: text/plain; charset="UTF-8"
Hello Folks,
I'm coming back to this after a few days of prodding etc. and there has
been some progress. The fault has been tracked down to the A3 input board
and it points at a crack in the board or a joint. Here is the sketch:
After a good clean of the switches and re-seating all boards etc. the 5370
sprung back into life with almost-in-spec performance of 102.xx ns with
it's own 10 MHz. Trouble is it didn't last - after a while and triggered by
sliding some switches around it went back to the 14 ns. Fiddling with the
BNC's may have the same effect, i.e. there is a mechanical element to it
all.
Turns out that running it with the front panel removed I can make it go
from 14 to 102 ns by slightly bending the A3 board, certainly while it is
cold. After a while this trick doesn't work anymore, my suspicion is
whatever crack/gap is causing the trouble has expanded too far to close it
tight. After cool-down we're back to square one.
I notice there are a few SMD components on that board, right in the middle
where it would bend most - basically on the reverse side of the switches.
Capacitors and resistors I expect. I wonder if it is worth going over the
solder joints of all these carefully.
What has stopped me so far are 2 questions:
Am I the only one with this observation or has anybody come across this
before?
Secondly the solder on these SMD's is coated with a black substance.
Clearly this could be removed somehow but it is probably there for a
reason. Has anybody re-touched joints like these before?
On an instrument like this it is very much a case of "proceed with care"
and I'd hate to do more damage by rushing in.
On a related subject: last year there was a discussion over redesigning the
input board(s) for the 5370 / 5345. Did this get off the ground? If my A3
packs up completely...
Thanks in advance for any suggestions,
Thomas.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
That is very good information. I will add your email to the 5370 page on my
web site.
Didier KO4BB
On Jul 18, 2017 4:02 AM, "Thomas Allgeier" th.allgeier@gmail.com wrote:
Hello All Again,
I’ve got my 5370B going now and in the process made a “discovery” which I
thought might be worth sharing:
The A3 input board is a through-hole PCB with a few SMD capacitors and
resistors on the reverse of the “switch area”. It turns out that on my 5370
(2410A00777) these components are not soldered, but fitted with conductive
adhesive. I first thought it was solder with a black coating but under a
microscope it is clear that it is not solder at all. Most probably it is a
mixture of epoxy and silver particles, or a similar compound. So no going
over joints with a fine iron…
Inspecting all this carefully under the microscope I discovered that the
“joints” on 2 resistors (R23 and R56) had cracked. As you know this board
gets heat from the hybrid amplifier IC’s and due to the way the board is
mounted to the front panel I guess it sees thermal stressing when the
instrument warms up and cools down. While this obviously lasts a long time
it looks that on my unit the adhesive has eventually cracked in places.
(Vigorous switch activation and pressing / pulling on the switch handles
also won’t be helpful in this respect…) One of the resistors just fell off
at the slightest touch with fine tweezers.
Anyhow, after removing the offending components, cleaning the pads of the
adhesive, and soldering replacements in place, we have a perfect 99.9x ns
with the 10 MHz on the commoned inputs. Happy days!
So if any of your 5370’s have the kind of intermittent fault I described
(and one or two other people seem to have reported) or instability that
seems to originate from the A3 board – check the joints around the SMD’s.
I wonder why / how it ended up having the adhesive instead of solder –
were earlier / later instruments the same, or was this a build change
introduced at a certain period?
Hope the above is of help to somebody else,
Thomas.
Message: 1
Date: Sat, 15 Jul 2017 21:54:47 +0100
From: Thomas Allgeier th.allgeier@gmail.com
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Subject: Re: [time-nuts] 5370B Question / help needed
Message-ID:
CAOT1cQLOjZ0=Xs_+nQ+MzAODboZG5ar0f8G_PydxJ=Oj6grB8g@mail.gmail.com
Content-Type: text/plain; charset="UTF-8"
Hello Folks,
I'm coming back to this after a few days of prodding etc. and there has
been some progress. The fault has been tracked down to the A3 input board
and it points at a crack in the board or a joint. Here is the sketch:
After a good clean of the switches and re-seating all boards etc. the 5370
sprung back into life with almost-in-spec performance of 102.xx ns with
it's own 10 MHz. Trouble is it didn't last - after a while and triggered by
sliding some switches around it went back to the 14 ns. Fiddling with the
BNC's may have the same effect, i.e. there is a mechanical element to it
all.
Turns out that running it with the front panel removed I can make it go
from 14 to 102 ns by slightly bending the A3 board, certainly while it is
cold. After a while this trick doesn't work anymore, my suspicion is
whatever crack/gap is causing the trouble has expanded too far to close it
tight. After cool-down we're back to square one.
I notice there are a few SMD components on that board, right in the middle
where it would bend most - basically on the reverse side of the switches.
Capacitors and resistors I expect. I wonder if it is worth going over the
solder joints of all these carefully.
What has stopped me so far are 2 questions:
Am I the only one with this observation or has anybody come across this
before?
Secondly the solder on these SMD's is coated with a black substance.
Clearly this could be removed somehow but it is probably there for a
reason. Has anybody re-touched joints like these before?
On an instrument like this it is very much a case of "proceed with care"
and I'd hate to do more damage by rushing in.
On a related subject: last year there was a discussion over redesigning the
input board(s) for the 5370 / 5345. Did this get off the ground? If my A3
packs up completely...
Thanks in advance for any suggestions,
Thomas.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.
Could it be that someone thought adhesive would be less prone to
cracking from stress than solder?
Also, do you think that the adhesive could be cleaned one end at a time
and replaced with solder?
I'm admittedly too lazy to look/try just now.
Don
On 2017-07-19 07:09, Gary Neilson wrote:
This is very interesting, I have a 5370B that has the same behavior as
yours. I will take the input board out again and give it a good
inspection. BTW, what did you use to clean the adhesive from the pads
?
Thanks
Gary
On 7/18/2017 2:41 AM, Thomas Allgeier wrote:
Hello All Again,
I’ve got my 5370B going now and in the process made a “discovery”
which I thought might be worth sharing:
The A3 input board is a through-hole PCB with a few SMD capacitors and
resistors on the reverse of the “switch area”. It turns out that on my
5370 (2410A00777) these components are not soldered, but fitted with
conductive adhesive. I first thought it was solder with a black
coating but under a microscope it is clear that it is not solder at
all. Most probably it is a mixture of epoxy and silver particles, or a
similar compound. So no going over joints with a fine iron…
Inspecting all this carefully under the microscope I discovered that
the “joints” on 2 resistors (R23 and R56) had cracked. As you know
this board gets heat from the hybrid amplifier IC’s and due to the way
the board is mounted to the front panel I guess it sees thermal
stressing when the instrument warms up and cools down. While this
obviously lasts a long time it looks that on my unit the adhesive has
eventually cracked in places. (Vigorous switch activation and pressing
/ pulling on the switch handles also won’t be helpful in this
respect…) One of the resistors just fell off at the slightest touch
with fine tweezers.
Anyhow, after removing the offending components, cleaning the pads of
the adhesive, and soldering replacements in place, we have a perfect
99.9x ns with the 10 MHz on the commoned inputs. Happy days!
So if any of your 5370’s have the kind of intermittent fault I
described (and one or two other people seem to have reported) or
instability that seems to originate from the A3 board – check the
joints around the SMD’s.
I wonder why / how it ended up having the adhesive instead of solder –
were earlier / later instruments the same, or was this a build change
introduced at a certain period?
Hope the above is of help to somebody else,
Thomas.
Message: 1
Date: Sat, 15 Jul 2017 21:54:47 +0100
From: Thomas Allgeier th.allgeier@gmail.com
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Subject: Re: [time-nuts] 5370B Question / help needed
Message-ID:
CAOT1cQLOjZ0=Xs_+nQ+MzAODboZG5ar0f8G_PydxJ=Oj6grB8g@mail.gmail.com
Content-Type: text/plain; charset="UTF-8"
Hello Folks,
I'm coming back to this after a few days of prodding etc. and there
has
been some progress. The fault has been tracked down to the A3 input
board
and it points at a crack in the board or a joint. Here is the sketch:
After a good clean of the switches and re-seating all boards etc. the
5370
sprung back into life with almost-in-spec performance of 102.xx ns
with
it's own 10 MHz. Trouble is it didn't last - after a while and
triggered by
sliding some switches around it went back to the 14 ns. Fiddling with
the
BNC's may have the same effect, i.e. there is a mechanical element to
it
all.
Turns out that running it with the front panel removed I can make it
go
from 14 to 102 ns by slightly bending the A3 board, certainly while it
is
cold. After a while this trick doesn't work anymore, my suspicion is
whatever crack/gap is causing the trouble has expanded too far to
close it
tight. After cool-down we're back to square one.
I notice there are a few SMD components on that board, right in the
middle
where it would bend most - basically on the reverse side of the
switches.
Capacitors and resistors I expect. I wonder if it is worth going over
the
solder joints of all these carefully.
What has stopped me so far are 2 questions:
Am I the only one with this observation or has anybody come across
this
before?
Secondly the solder on these SMD's is coated with a black substance.
Clearly this could be removed somehow but it is probably there for a
reason. Has anybody re-touched joints like these before?
On an instrument like this it is very much a case of "proceed with
care"
and I'd hate to do more damage by rushing in.
On a related subject: last year there was a discussion over
redesigning the
input board(s) for the 5370 / 5345. Did this get off the ground? If my
A3
packs up completely...
Thanks in advance for any suggestions,
Thomas.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
--
Dr. Don Latham
PO Box 404, Frenchtown, MT, 59834
VOX: 406-626-4304
Thanks! Didier! 73, John K1AE
-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Didier Juges
Sent: Wednesday, July 19, 2017 1:35 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] 5370B Input Board Repair [WAS: 5370B Question / help needed]
That is very good information. I will add your email to the 5370 page on my
web site.
Didier KO4BB
On Jul 18, 2017 4:02 AM, "Thomas Allgeier" th.allgeier@gmail.com wrote:
Hello All Again,
I’ve got my 5370B going now and in the process made a “discovery” which I
thought might be worth sharing:
The A3 input board is a through-hole PCB with a few SMD capacitors and
resistors on the reverse of the “switch area”. It turns out that on my 5370
(2410A00777) these components are not soldered, but fitted with conductive
adhesive. I first thought it was solder with a black coating but under a
microscope it is clear that it is not solder at all. Most probably it is a
mixture of epoxy and silver particles, or a similar compound. So no going
over joints with a fine iron…
Inspecting all this carefully under the microscope I discovered that the
“joints” on 2 resistors (R23 and R56) had cracked. As you know this board
gets heat from the hybrid amplifier IC’s and due to the way the board is
mounted to the front panel I guess it sees thermal stressing when the
instrument warms up and cools down. While this obviously lasts a long time
it looks that on my unit the adhesive has eventually cracked in places.
(Vigorous switch activation and pressing / pulling on the switch handles
also won’t be helpful in this respect…) One of the resistors just fell off
at the slightest touch with fine tweezers.
Anyhow, after removing the offending components, cleaning the pads of the
adhesive, and soldering replacements in place, we have a perfect 99.9x ns
with the 10 MHz on the commoned inputs. Happy days!
So if any of your 5370’s have the kind of intermittent fault I described
(and one or two other people seem to have reported) or instability that
seems to originate from the A3 board – check the joints around the SMD’s.
I wonder why / how it ended up having the adhesive instead of solder –
were earlier / later instruments the same, or was this a build change
introduced at a certain period?
Hope the above is of help to somebody else,
Thomas.
Message: 1
Date: Sat, 15 Jul 2017 21:54:47 +0100
From: Thomas Allgeier th.allgeier@gmail.com
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Subject: Re: [time-nuts] 5370B Question / help needed
Message-ID:
CAOT1cQLOjZ0=Xs_+nQ+MzAODboZG5ar0f8G_PydxJ=Oj6grB8g@mail.gmail.com
Content-Type: text/plain; charset="UTF-8"
Hello Folks,
I'm coming back to this after a few days of prodding etc. and there has
been some progress. The fault has been tracked down to the A3 input board
and it points at a crack in the board or a joint. Here is the sketch:
After a good clean of the switches and re-seating all boards etc. the 5370
sprung back into life with almost-in-spec performance of 102.xx ns with
it's own 10 MHz. Trouble is it didn't last - after a while and triggered by
sliding some switches around it went back to the 14 ns. Fiddling with the
BNC's may have the same effect, i.e. there is a mechanical element to it
all.
Turns out that running it with the front panel removed I can make it go
from 14 to 102 ns by slightly bending the A3 board, certainly while it is
cold. After a while this trick doesn't work anymore, my suspicion is
whatever crack/gap is causing the trouble has expanded too far to close it
tight. After cool-down we're back to square one.
I notice there are a few SMD components on that board, right in the middle
where it would bend most - basically on the reverse side of the switches.
Capacitors and resistors I expect. I wonder if it is worth going over the
solder joints of all these carefully.
What has stopped me so far are 2 questions:
Am I the only one with this observation or has anybody come across this
before?
Secondly the solder on these SMD's is coated with a black substance.
Clearly this could be removed somehow but it is probably there for a
reason. Has anybody re-touched joints like these before?
On an instrument like this it is very much a case of "proceed with care"
and I'd hate to do more damage by rushing in.
On a related subject: last year there was a discussion over redesigning the
input board(s) for the 5370 / 5345. Did this get off the ground? If my A3
packs up completely...
Thanks in advance for any suggestions,
Thomas.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
This email has been checked for viruses by Avast antivirus software.