SR620/PM66xx/CNT-90 input stages

AK
Attila Kinali
Tue, Jun 20, 2017 7:20 AM

Hi,

I had a look at the PM668x[1] and CNT-90[2] schematics yesterday and
noticed one thing: The input stages are strikingly similar to the
SR620[3] (down to the parts used) and all of them have a gain 1 amplifier
infront of the comparator. The PM668x service manual explicitly calls it
"impedance converter", even. Now, if the downstream circuit would be low
impedance, I could understand that, but the sink is a comparator with a
high impedance input (only a few µA input current). I am sure the engineers
had a good reason to add those amplifiers, but I cannot guess why. Would
someone be so kind and enlighten me?

		Attila Kinali

[1] http://bee.mif.pg.gda.pl/ciasteczkowypotwor/Fluke/Fluke_PM6681_Service_Manual.pdf
[2] http://assets.fluke.com/manuals/6690____smeng0000.pdf
[3] http://www.ko4bb.com/getsimple/index.php?id=download&file=02_GPS_Timing/Stanford_Research_Systems/SR620_Universal_Time_Interval_Counter_Schematics.pdf

--
It is upon moral qualities that a society is ultimately founded. All
the prosperity and technological sophistication in the world is of no
use without that foundation.
-- Miss Matheson, The Diamond Age, Neil Stephenson

Hi, I had a look at the PM668x[1] and CNT-90[2] schematics yesterday and noticed one thing: The input stages are strikingly similar to the SR620[3] (down to the parts used) and all of them have a gain 1 amplifier infront of the comparator. The PM668x service manual explicitly calls it "impedance converter", even. Now, if the downstream circuit would be low impedance, I could understand that, but the sink is a comparator with a high impedance input (only a few µA input current). I am sure the engineers had a good reason to add those amplifiers, but I cannot guess why. Would someone be so kind and enlighten me? Attila Kinali [1] http://bee.mif.pg.gda.pl/ciasteczkowypotwor/Fluke/Fluke_PM6681_Service_Manual.pdf [2] http://assets.fluke.com/manuals/6690____smeng0000.pdf [3] http://www.ko4bb.com/getsimple/index.php?id=download&file=02_GPS_Timing/Stanford_Research_Systems/SR620_Universal_Time_Interval_Counter_Schematics.pdf -- It is upon moral qualities that a society is ultimately founded. All the prosperity and technological sophistication in the world is of no use without that foundation. -- Miss Matheson, The Diamond Age, Neil Stephenson
BG
Bruce Griffiths
Tue, Jun 20, 2017 9:57 AM

Its somewhat difficult to achieve a 1Megohm input with just a bipolar comparator input stage especially if AC coupling is required.

Bruce

 On 20 June 2017 at 19:20 Attila Kinali <attila@kinali.ch> wrote:

 Hi,

 I had a look at the PM668x[1] and CNT-90[2] schematics yesterday and
 noticed one thing: The input stages are strikingly similar to the
 SR620[3] (down to the parts used) and all of them have a gain 1 amplifier
 infront of the comparator. The PM668x service manual explicitly calls it
 "impedance converter", even. Now, if the downstream circuit would be low
 impedance, I could understand that, but the sink is a comparator with a
 high impedance input (only a few µA input current). I am sure the engineers
 had a good reason to add those amplifiers, but I cannot guess why. Would
 someone be so kind and enlighten me?

 Attila Kinali

 [1] http://bee.mif.pg.gda.pl/ciasteczkowypotwor/Fluke/Fluke_PM6681_Service_Manual.pdf
 [2] http://assets.fluke.com/manuals/6690____smeng0000.pdf
 [3] http://www.ko4bb.com/getsimple/index.php?id=download&file=02_GPS_Timing/Stanford_Research_Systems/SR620_Universal_Time_Interval_Counter_Schematics.pdf http://www.ko4bb.com/getsimple/index.php?id=download&file=02_GPS_Timing/Stanford_Research_Systems/SR620_Universal_Time_Interval_Counter_Schematics.pdf

 --
 It is upon moral qualities that a society is ultimately founded. All
 the prosperity and technological sophistication in the world is of no
 use without that foundation.
 -- Miss Matheson, The Diamond Age, Neil Stephenson

 _______________________________________________
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
Its somewhat difficult to achieve a 1Megohm input with just a bipolar comparator input stage especially if AC coupling is required. Bruce > > On 20 June 2017 at 19:20 Attila Kinali <attila@kinali.ch> wrote: > > Hi, > > I had a look at the PM668x[1] and CNT-90[2] schematics yesterday and > noticed one thing: The input stages are strikingly similar to the > SR620[3] (down to the parts used) and all of them have a gain 1 amplifier > infront of the comparator. The PM668x service manual explicitly calls it > "impedance converter", even. Now, if the downstream circuit would be low > impedance, I could understand that, but the sink is a comparator with a > high impedance input (only a few µA input current). I am sure the engineers > had a good reason to add those amplifiers, but I cannot guess why. Would > someone be so kind and enlighten me? > > Attila Kinali > > [1] http://bee.mif.pg.gda.pl/ciasteczkowypotwor/Fluke/Fluke_PM6681_Service_Manual.pdf > [2] http://assets.fluke.com/manuals/6690____smeng0000.pdf > [3] http://www.ko4bb.com/getsimple/index.php?id=download&file=02_GPS_Timing/Stanford_Research_Systems/SR620_Universal_Time_Interval_Counter_Schematics.pdf http://www.ko4bb.com/getsimple/index.php?id=download&file=02_GPS_Timing/Stanford_Research_Systems/SR620_Universal_Time_Interval_Counter_Schematics.pdf > > -- > It is upon moral qualities that a society is ultimately founded. All > the prosperity and technological sophistication in the world is of no > use without that foundation. > -- Miss Matheson, The Diamond Age, Neil Stephenson > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
LA
Li Ang
Tue, Jun 20, 2017 10:01 AM

Hi
I think the purpose is to get high input impedance for high frequency signal. The input impedance of comparator is high at low frequency range only.

Li Ang / BI7LNQ
---Original---
From: "Attila Kinali"attila@kinali.ch
Date: 2017/6/20 15:20:44
To: "Discussion of precise time and frequency measurement"time-nuts@febo.com;
Subject: [time-nuts] SR620/PM66xx/CNT-90 input stages

Hi,

I had a look at the PM668x[1] and CNT-90[2] schematics yesterday and
noticed one thing: The input stages are strikingly similar to the
SR620[3] (down to the parts used) and all of them have a gain 1 amplifier
infront of the comparator. The PM668x service manual explicitly calls it
"impedance converter", even. Now, if the downstream circuit would be low
impedance, I could understand that, but the sink is a comparator with a
high impedance input (only a few µA input current). I am sure the engineers
had a good reason to add those amplifiers, but I cannot guess why. Would
someone be so kind and enlighten me?

Attila Kinali

[1] http://bee.mif.pg.gda.pl/ciasteczkowypotwor/Fluke/Fluke_PM6681_Service_Manual.pdf
[2] http://assets.fluke.com/manuals/6690____smeng0000.pdf
[3] http://www.ko4bb.com/getsimple/index.php?id=download&file=02_GPS_Timing/Stanford_Research_Systems/SR620_Universal_Time_Interval_Counter_Schematics.pdf

--
It is upon moral qualities that a society is ultimately founded. All
the prosperity and technological sophistication in the world is of no
use without that foundation.
-- Miss Matheson, The Diamond Age, Neil Stephenson


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi I think the purpose is to get high input impedance for high frequency signal. The input impedance of comparator is high at low frequency range only. Li Ang / BI7LNQ ---Original--- From: "Attila Kinali"<attila@kinali.ch> Date: 2017/6/20 15:20:44 To: "Discussion of precise time and frequency measurement"<time-nuts@febo.com>; Subject: [time-nuts] SR620/PM66xx/CNT-90 input stages Hi, I had a look at the PM668x[1] and CNT-90[2] schematics yesterday and noticed one thing: The input stages are strikingly similar to the SR620[3] (down to the parts used) and all of them have a gain 1 amplifier infront of the comparator. The PM668x service manual explicitly calls it "impedance converter", even. Now, if the downstream circuit would be low impedance, I could understand that, but the sink is a comparator with a high impedance input (only a few µA input current). I am sure the engineers had a good reason to add those amplifiers, but I cannot guess why. Would someone be so kind and enlighten me? Attila Kinali [1] http://bee.mif.pg.gda.pl/ciasteczkowypotwor/Fluke/Fluke_PM6681_Service_Manual.pdf [2] http://assets.fluke.com/manuals/6690____smeng0000.pdf [3] http://www.ko4bb.com/getsimple/index.php?id=download&file=02_GPS_Timing/Stanford_Research_Systems/SR620_Universal_Time_Interval_Counter_Schematics.pdf -- It is upon moral qualities that a society is ultimately founded. All the prosperity and technological sophistication in the world is of no use without that foundation. -- Miss Matheson, The Diamond Age, Neil Stephenson _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
CS
Charles Steinmetz
Tue, Jun 20, 2017 9:11 PM

Attila wrote:

Now, if the downstream circuit would be low
impedance, I could understand that, but the sink is a comparator with a
high impedance input (only a few µA input current). I am sure the engineers
had a good reason to add those amplifiers, but I cannot guess why. Would
someone be so kind and enlighten me?

I was going tp post the Smith chart for the input of the AD96885/7 ECL
comparator they use, but was surprised to find that AD did not include
it on the datasheet.  However, just from the spec table we can see why a
buffer is necessary:

Input Resistance:      200k ohms
Input Capacitance:    2pF

For starters, even at low frequencies the input impedance is only 200k
ohms, much less than the 1Mohm rated impedance.  Further, the
capacitance seriously degrades bandwidth to as low as 400kHz (this
depends on the source impedance, but even with a 10kohm source the BW is
only 8MHz).

(Note that the input current spec is not necessarily a good proxy for
input impedance -- consider a FET-input amplifier with an input current
of 1uA and a 100k gate resistor.)

Best regards,

Charles

Attila wrote: > Now, if the downstream circuit would be low > impedance, I could understand that, but the sink is a comparator with a > high impedance input (only a few µA input current). I am sure the engineers > had a good reason to add those amplifiers, but I cannot guess why. Would > someone be so kind and enlighten me? I was going tp post the Smith chart for the input of the AD96885/7 ECL comparator they use, but was surprised to find that AD did not include it on the datasheet. However, just from the spec table we can see why a buffer is necessary: Input Resistance: 200k ohms Input Capacitance: 2pF For starters, even at low frequencies the input impedance is only 200k ohms, much less than the 1Mohm rated impedance. Further, the capacitance seriously degrades bandwidth to as low as 400kHz (this depends on the source impedance, but even with a 10kohm source the BW is only 8MHz). (Note that the input current spec is not necessarily a good proxy for input impedance -- consider a FET-input amplifier with an input current of 1uA and a 100k gate resistor.) Best regards, Charles
BG
Bruce Griffiths
Tue, Jun 20, 2017 9:35 PM

If the source impendance at dc is 10k at dc a bias current of a few microamp produces a dc offset of tens of millivolts which may be an issue if the signal amplitude is low and has a low slew rate.

Bruce

 On 21 June 2017 at 09:11 Charles Steinmetz <csteinmetz@yandex.com> wrote:

 Attila wrote:
     Now, if the downstream circuit would be low
     impedance, I could understand that, but the sink is a comparator with a
     high impedance input (only a few µA input current). I am sure the engineers
     had a good reason to add those amplifiers, but I cannot guess why. Would
     someone be so kind and enlighten me?
 I was going tp post the Smith chart for the input of the AD96885/7 ECL
 comparator they use, but was surprised to find that AD did not include
 it on the datasheet. However, just from the spec table we can see why a
 buffer is necessary:

 Input Resistance: 200k ohms
 Input Capacitance: 2pF

 For starters, even at low frequencies the input impedance is only 200k
 ohms, much less than the 1Mohm rated impedance. Further, the
 capacitance seriously degrades bandwidth to as low as 400kHz (this
 depends on the source impedance, but even with a 10kohm source the BW is
 only 8MHz).

 (Note that the input current spec is not necessarily a good proxy for
 input impedance -- consider a FET-input amplifier with an input current
 of 1uA and a 100k gate resistor.)

 Best regards,

 Charles

 _______________________________________________
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
If the source impendance at dc is 10k at dc a bias current of a few microamp produces a dc offset of tens of millivolts which may be an issue if the signal amplitude is low and has a low slew rate. Bruce > > On 21 June 2017 at 09:11 Charles Steinmetz <csteinmetz@yandex.com> wrote: > > Attila wrote: > > > > > > Now, if the downstream circuit would be low > > impedance, I could understand that, but the sink is a comparator with a > > high impedance input (only a few µA input current). I am sure the engineers > > had a good reason to add those amplifiers, but I cannot guess why. Would > > someone be so kind and enlighten me? > > > > > > I was going tp post the Smith chart for the input of the AD96885/7 ECL > comparator they use, but was surprised to find that AD did not include > it on the datasheet. However, just from the spec table we can see why a > buffer is necessary: > > Input Resistance: 200k ohms > Input Capacitance: 2pF > > For starters, even at low frequencies the input impedance is only 200k > ohms, much less than the 1Mohm rated impedance. Further, the > capacitance seriously degrades bandwidth to as low as 400kHz (this > depends on the source impedance, but even with a 10kohm source the BW is > only 8MHz). > > (Note that the input current spec is not necessarily a good proxy for > input impedance -- consider a FET-input amplifier with an input current > of 1uA and a 100k gate resistor.) > > Best regards, > > Charles > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
AK
Attila Kinali
Wed, Jun 21, 2017 8:34 AM

Moin,

On Tue, 20 Jun 2017 17:11:24 -0400
Charles Steinmetz csteinmetz@yandex.com wrote:

I was going tp post the Smith chart for the input of the AD96885/7 ECL
comparator they use, but was surprised to find that AD did not include
it on the datasheet.  However, just from the spec table we can see why a
buffer is necessary:

Input Resistance:      200k ohms
Input Capacitance:    2pF

Thanks everyone! I was caught again in my DC line of thinking.
Even though I saw the capacitance spec, it didn't registers.

At some point I need to learn how to design RF circuits....

		Attila Kinali

--
It is upon moral qualities that a society is ultimately founded. All
the prosperity and technological sophistication in the world is of no
use without that foundation.
-- Miss Matheson, The Diamond Age, Neil Stephenson

Moin, On Tue, 20 Jun 2017 17:11:24 -0400 Charles Steinmetz <csteinmetz@yandex.com> wrote: > I was going tp post the Smith chart for the input of the AD96885/7 ECL > comparator they use, but was surprised to find that AD did not include > it on the datasheet. However, just from the spec table we can see why a > buffer is necessary: > > Input Resistance: 200k ohms > Input Capacitance: 2pF Thanks everyone! I was caught again in my DC line of thinking. Even though I saw the capacitance spec, it didn't registers. At some point I need to learn how to design RF circuits.... Attila Kinali -- It is upon moral qualities that a society is ultimately founded. All the prosperity and technological sophistication in the world is of no use without that foundation. -- Miss Matheson, The Diamond Age, Neil Stephenson
CS
Charles Steinmetz
Wed, Jun 21, 2017 10:14 AM

Bruce wrote:

If the source impendance at dc is 10k at dc a bias current of a few
microamp produces a dc offset of tens of millivolts which may be an
issue if the signal amplitude is low and has a low slew rate.

Jeez, Bruce, it was nothing but a thought experiment to illustrate why
an amplifier's DC input current is not a reliable proxy for input
impedance  It was not a suggestion for how to design an amplifier input.
And that should have been obvious to you -- I'm sure it was to
everyone else.

Get a life.

Charles

Bruce wrote: > If the source impendance at dc is 10k at dc a bias current of a few > microamp produces a dc offset of tens of millivolts which may be an > issue if the signal amplitude is low and has a low slew rate. Jeez, Bruce, it was nothing but a thought experiment to illustrate why an amplifier's DC input current is not a reliable proxy for input impedance It was not a suggestion for how to design an amplifier input. And that should have been obvious to you -- I'm sure it was to everyone else. Get a life. Charles
BG
Bruce Griffiths
Wed, Jun 21, 2017 10:40 AM

Your assumptions about my comment are unfounded. It was just intended as an additional reason as to why direct connection to a comparator input isnt a good idea with a high source impedance. The original post asked why a buffer was needed.

Bruce

 On 21 June 2017 at 22:14 Charles Steinmetz <csteinmetz@yandex.com> wrote:

 Bruce wrote:
     If the source impendance at dc is 10k at dc a bias current of a few
     microamp produces a dc offset of tens of millivolts which may be an
     issue if the signal amplitude is low and has a low slew rate.
 Jeez, Bruce, it was nothing but a thought experiment to illustrate why
 an amplifier's DC input current is not a reliable proxy for input
 impedance It was not a suggestion for how to design an amplifier input.
 And that should have been obvious to you -- I'm sure it was to
 everyone else.

 Get a life.

 Charles

 _______________________________________________
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
Your assumptions about my comment are unfounded. It was just intended as an additional reason as to why direct connection to a comparator input isnt a good idea with a high source impedance. The original post asked why a buffer was needed. Bruce > > On 21 June 2017 at 22:14 Charles Steinmetz <csteinmetz@yandex.com> wrote: > > Bruce wrote: > > > > > > If the source impendance at dc is 10k at dc a bias current of a few > > microamp produces a dc offset of tens of millivolts which may be an > > issue if the signal amplitude is low and has a low slew rate. > > > > > > Jeez, Bruce, it was nothing but a thought experiment to illustrate why > an amplifier's DC input current is not a reliable proxy for input > impedance It was not a suggestion for how to design an amplifier input. > And that should have been obvious to you -- I'm sure it was to > everyone else. > > Get a life. > > Charles > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >