A3 ADC has no way to measure meter's reference drift, as all of the measurements are relative to A9 output and/or 40k STDR. In this respect ref output is taken as constant. As result drifty reference will cause all readings to drift as well after self-calibration, because DC constant is changed only after external DCV 10V cal. So you have either zero A9 drift assumed from ACAL DCV or zero drift ADC A3 assumed from CAL 10V. By playing with time you can narrow the more drifty component. Good stable 3458A stable to <0.2ppm over week.
On November 24, 2017 9:32:30 AM GMT+08:00, Randy Evans randyevans2688@gmail.com wrote:
Illya,
Can you explain why you say " If ACAL DCV does not remove drift then A3
is
probably fine". I don't really follow the argument.
Randy
On Thu, Nov 23, 2017 at 5:18 PM, Illya Tsemenko illya@xdevs.com
wrote:
Well, one thing you can know for sure , that reference is indeed
broken.
Drift over a day should be way below the noise floor (<0.1ppm). With
1.7x
gain of ADC scale that drift rate gives you around 0.8ppm +/- 0.3ppm
due to
zener noise. So it is in line of 1.1ppm/day. If ACAL DCV does not
remove
drift then A3 is probably fine.
Since reference is much easier to troubleshoot and fix I would go
with
replacing LTZ chip and testing if drift go away, if that have any
help on
your desire to keep meter.
Also serial number range in SN doesn't mean much for you, as meter
history
is unknown and it still may have been serviced at some time.
On November 24, 2017 8:40:00 AM GMT+08:00, Randy Evans <
randyevans2688@gmail.com> wrote:
Illya,
I ran the test you suggested over 4 days and got 7.19114068 VDC on
the
start of the test and 7.19113736 VDC at the end of the fourth day
(96 hours
later). I calculate a total of 0.46 ppm drift, which seems
excessive but
does not account for the 1.1 ppm/day I measured overall. I suspect
the A3
card is the primary source of drifting. Since the unit is a late
model
Agilent unit, that is well beyond the expected range of units
described in
Service Note 18. What do you think?
Thanks,
Randy Evans
On Sat, Nov 18, 2017 at 4:47 AM, Illya Tsemenko illya@xdevs.com
wrote:
Since you have 732A, testing should be easy enough. Calibrate
faulty
meter for zero and DCV 10V to 732A, record CAL? 2,1 value. This is
your LTZ
output. Then leave it running for few days to drift away and
calibrate
again to same 732A. Check CAL? 2,1 again. Calculate the difference
and if
it matches output drift (that 1.1ppm/day you mention) - you can be
80% sure
that A9 is a problem. Other 19% go to A1 and A3 circuits, as 7V is
not used
directly in the meter, and there are still gain parts to get +12
and
-12VREF on A3 and 10Vish bipolar levels on A1. If your CAL? 2,1
stays same
(within 0.3ppm) then A9 is fine.
On November 18, 2017 12:59:53 PM GMT+08:00, Randy Evans <
randyevans2688@gmail.com> wrote:
I just received an Agilent 3458A that has a problem with noise and
a
drifting voltage measurements. I am using two Fluke 732As to
compare
absolute voltage measurements over time against the Agilent and an
HP
3458A. The HP unit has a new A3 ADC card and seems to be very
stable and
low noise, so is being used for comparison. I have been doing
simultaneous
absolute voltage measurements and DC Cal Constant measurements
several
times a day and then calculating the drift rates of the two units
using the
HP Service Note 18 procedure.
The results indicate the Cal Constant drift rate of both units is
very
similar and within spec per Service Note 18. However, the
absolute value
measurements show the Agilent unit changing 1.1 ppm over a day
whereas the
HP unit is within a tenth of a ppm over a day. In my way of
thinking the
Cal Constant procedure assumes the voltage reference board in the
3458A is
stable, hence the absolute value reading should remain essentially
constant
after each ACAL DCV, which is the case with the HP unit. Since
the Agilent
unit shows a steady drift in the absolute reading, this would
indicate to
me that the voltage reference board is likely the cause of the
problem, and
is also likely the cause of the noisy readings. If so, this is a
“relatively” easy fix (I have several 3458A voltage reference
boards, one
of which has been continuously powered up for several years).
The issue is that I have to make a decision to keep or return the
Agilent.
It has a cal seal on it and if I open the unit up to change the
voltage
reference board, I own it and can’t return it. I would appreciate
an
opinion from the members of the group as to what they think the
odds are
that the voltage reference board is the source of the problems
with the
Agilent 3458A.
Thanks,
Randy Evans