After running an ACAL 0 with a four wire short in place I get a DCV reading
starting at about -0.00005 mV to increasing to about 0.00060 to 0.00070 mV
over a minute or so at which point it seems to stabilise.
Is this "normal" and if not what could be wrong (AZERO is turned on).
Dave
In message 003801d389fb$bbba1f10$332e5d30$@perdrix.co.uk, "David C. Partridge
" writes:
After running an ACAL 0 with a four wire short in place I get a DCV reading
starting at about -0.00005 mV to increasing to about 0.00060 to 0.00070 mV
over a minute or so at which point it seems to stabilise.
That is pretty normal, it's very hard to control temperatures so precisely
that you don't get a few hundred nanovolts of seebeck voltage.
If you touch the short with your finger, you will change the temperature
enough that it registers clearly.
I mount the short, then hang a piece of cloth over the terminals and wait
about 15-30 minutes before I run ACAL 0
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
Correction to -0.00060mV to -0.00070mV
Dave
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of David C. Partridge
Sent: 10 January 2018 10:14
To: 'Discussion of precise voltage measurement'
Subject: [volt-nuts] 3458A question
After running an ACAL 0 with a four wire short in place I get a DCV reading starting at about -0.00005 mV to increasing to about 0.00060 to 0.00070 mV over a minute or so at which point it seems to stabilise.
Is this "normal" and if not what could be wrong (AZERO is turned on).
Dave
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
To add some more information
The meter has now been on for a number of days with a newly cleaned 4 wire short in place all the time.
I have run a full CAL 0 to zero the offsets, but even so, and after running an ACAL 0, I get a DCV reading of -200 to -400nV (0.00040mV).
Is that normal? I'd have thought it would typically be less than that given that the 1 year limit is 0.00106mV (1.06Uv).
Thanks again
David
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of David C. Partridge
Sent: 10 January 2018 10:25
To: 'Discussion of precise voltage measurement'
Subject: Re: [volt-nuts] 3458A question
Correction to -0.00060mV to -0.00070mV
Dave
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of David C. Partridge
Sent: 10 January 2018 10:14
To: 'Discussion of precise voltage measurement'
Subject: [volt-nuts] 3458A question
After running an ACAL 0 with a four wire short in place I get a DCV reading starting at about -0.00005 mV to increasing to about 0.00060 to 0.00070 mV over a minute or so at which point it seems to stabilise.
Is this "normal" and if not what could be wrong (AZERO is turned on).
Dave
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
In message 000001d38ae3$ea81a2d0$bf84e870$@perdrix.co.uk, "David C. Partridge
" writes:
The meter has now been on for a number of days with a newly cleaned 4 wire short in place all the time.
I have run a full CAL 0 to zero the offsets, but even so, and after running an ACAL 0, I get a DCV reading of -200 to -400nV (0.00040mV).
Is that normal?
I see the same kind of numbers.
--