I hope to have a 7081 in the next few weeks that is minus the test leads and
calibration key.  I have been able to chase down the connector for the front
panel test leads and two are on the way.  However, in fabricating some test
leads, I was wondering if there is anything special about the wires/cable
used to make up the test leads.  Silver wire?  Teflon insulation?  Etc.?

I was planning on using 18 ga. Teflon insulated stranded wire used in
aircraft and some standard 'probes'.  For the Kelvin leads, I was thinking
about using alligator clips.  Any thoughts, insight, suggestions, previous
experience, etc., would be appreciated.

Thanks in advance.

Joe

I hope to have a 7081 in the next few weeks that is minus the test leads and calibration key. I have been able to chase down the connector for the front panel test leads and two are on the way. However, in fabricating some test leads, I was wondering if there is anything special about the wires/cable used to make up the test leads. Silver wire? Teflon insulation? Etc.? I was planning on using 18 ga. Teflon insulated stranded wire used in aircraft and some standard 'probes'. For the Kelvin leads, I was thinking about using alligator clips. Any thoughts, insight, suggestions, previous experience, etc., would be appreciated. Thanks in advance. Joe
MK
m k
Thu, Aug 4, 2011 6:11 AM

Hi Joe,

You will need tellurium copper banana plugs to control the thermal emf's at the meter end and consider the materials that the alligator clips are made of for unintended thermocouples with the copper leads, teflon is a good insulation to have.

Mike

From: jltran@att.net
To: volt-nuts@febo.com
Date: Wed, 3 Aug 2011 17:25:07 -0500
Subject: [volt-nuts] Solartron 7081 Test Leads

I hope to have a 7081 in the next few weeks that is minus the test leads and
calibration key.  I have been able to chase down the connector for the front
panel test leads and two are on the way.  However, in fabricating some test
leads, I was wondering if there is anything special about the wires/cable
used to make up the test leads.  Silver wire?  Teflon insulation?  Etc.?

I was planning on using 18 ga. Teflon insulated stranded wire used in
aircraft and some standard 'probes'.  For the Kelvin leads, I was thinking
about using alligator clips.  Any thoughts, insight, suggestions, previous
experience, etc., would be appreciated.

Thanks in advance.

Joe


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi Joe, You will need tellurium copper banana plugs to control the thermal emf's at the meter end and consider the materials that the alligator clips are made of for unintended thermocouples with the copper leads, teflon is a good insulation to have. Mike > From: jltran@att.net > To: volt-nuts@febo.com > Date: Wed, 3 Aug 2011 17:25:07 -0500 > Subject: [volt-nuts] Solartron 7081 Test Leads > > I hope to have a 7081 in the next few weeks that is minus the test leads and > calibration key. I have been able to chase down the connector for the front > panel test leads and two are on the way. However, in fabricating some test > leads, I was wondering if there is anything special about the wires/cable > used to make up the test leads. Silver wire? Teflon insulation? Etc.? > > > > I was planning on using 18 ga. Teflon insulated stranded wire used in > aircraft and some standard 'probes'. For the Kelvin leads, I was thinking > about using alligator clips. Any thoughts, insight, suggestions, previous > experience, etc., would be appreciated. > > > > Thanks in advance. > > > > Joe > > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
FS
Fred Schneider
Thu, Aug 4, 2011 6:16 AM

Where did you find the connector, I have a 7060 and It would be nice to have a second cable.

Fred PA4TIM

Op 4 aug. 2011 om 08:11 heeft m k m1k3k1@hotmail.com het volgende geschreven:

Hi Joe,

You will need tellurium copper banana plugs to control the thermal emf's at the meter end and consider the materials that the alligator clips are made of for unintended thermocouples with the copper leads, teflon is a good insulation to have.

Mike

From: jltran@att.net
To: volt-nuts@febo.com
Date: Wed, 3 Aug 2011 17:25:07 -0500
Subject: [volt-nuts] Solartron 7081 Test Leads

I hope to have a 7081 in the next few weeks that is minus the test leads and
calibration key.  I have been able to chase down the connector for the front
panel test leads and two are on the way.  However, in fabricating some test
leads, I was wondering if there is anything special about the wires/cable
used to make up the test leads.  Silver wire?  Teflon insulation?  Etc.?

I was planning on using 18 ga. Teflon insulated stranded wire used in
aircraft and some standard 'probes'.  For the Kelvin leads, I was thinking
about using alligator clips.  Any thoughts, insight, suggestions, previous
experience, etc., would be appreciated.

Thanks in advance.

Joe


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Where did you find the connector, I have a 7060 and It would be nice to have a second cable. Fred PA4TIM Op 4 aug. 2011 om 08:11 heeft m k <m1k3k1@hotmail.com> het volgende geschreven: > > Hi Joe, > > You will need tellurium copper banana plugs to control the thermal emf's at the meter end and consider the materials that the alligator clips are made of for unintended thermocouples with the copper leads, teflon is a good insulation to have. > > Mike > >> From: jltran@att.net >> To: volt-nuts@febo.com >> Date: Wed, 3 Aug 2011 17:25:07 -0500 >> Subject: [volt-nuts] Solartron 7081 Test Leads >> >> I hope to have a 7081 in the next few weeks that is minus the test leads and >> calibration key. I have been able to chase down the connector for the front >> panel test leads and two are on the way. However, in fabricating some test >> leads, I was wondering if there is anything special about the wires/cable >> used to make up the test leads. Silver wire? Teflon insulation? Etc.? >> >> >> >> I was planning on using 18 ga. Teflon insulated stranded wire used in >> aircraft and some standard 'probes'. For the Kelvin leads, I was thinking >> about using alligator clips. Any thoughts, insight, suggestions, previous >> experience, etc., would be appreciated. >> >> >> >> Thanks in advance. >> >> >> >> Joe >> >> >> >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
DC
David C. Partridge
Thu, Aug 4, 2011 9:22 AM

And where to purchase TeCu banana plugs? Enquiring minds ...

The only source I've ever found was the Eichmann ETI which is made for the audio market, and the price is about USD90 for four

Thanks
Dave
-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of m k
Sent: 04 August 2011 07:11
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Solartron 7081 Test Leads

Hi Joe,

You will need tellurium copper banana plugs

And where to purchase TeCu banana plugs? Enquiring minds ... The only source I've ever found was the Eichmann ETI which is made for the audio market, and the price is about USD90 for four Thanks Dave -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of m k Sent: 04 August 2011 07:11 To: volt-nuts@febo.com Subject: Re: [volt-nuts] Solartron 7081 Test Leads Hi Joe, You will need tellurium copper banana plugs
JL
J. L. Trantham
Thu, Aug 4, 2011 11:50 AM

Mike,

I am new to the arena of high accuracy measurements but the plan is for this
to be the 'lab standard'.

There are no banana sockets on the front of the 7081.  It is a Fischer 5 pin
connector that took a while to chase down the part numbers.  I wanted to be
able to find the correct connector to mate with the unit to preserve its
full level of functionality.

Any suggestions of where to look for probes and alligator clips?

Now to make up some cables.  Alternatively, are cables available from
Solartron or another supplier?

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of m k
Sent: Thursday, August 04, 2011 1:11 AM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Solartron 7081 Test Leads

Hi Joe,

You will need tellurium copper banana plugs to control the thermal emf's at
the meter end and consider the materials that the alligator clips are made
of for unintended thermocouples with the copper leads, teflon is a good
insulation to have.

Mike

From: jltran@att.net
To: volt-nuts@febo.com
Date: Wed, 3 Aug 2011 17:25:07 -0500
Subject: [volt-nuts] Solartron 7081 Test Leads

I hope to have a 7081 in the next few weeks that is minus the test leads

and

calibration key.  I have been able to chase down the connector for the

front

panel test leads and two are on the way.  However, in fabricating some

test

leads, I was wondering if there is anything special about the wires/cable
used to make up the test leads.  Silver wire?  Teflon insulation?  Etc.?

I was planning on using 18 ga. Teflon insulated stranded wire used in
aircraft and some standard 'probes'.  For the Kelvin leads, I was thinking
about using alligator clips.  Any thoughts, insight, suggestions, previous
experience, etc., would be appreciated.

Thanks in advance.

Joe


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Mike, I am new to the arena of high accuracy measurements but the plan is for this to be the 'lab standard'. There are no banana sockets on the front of the 7081. It is a Fischer 5 pin connector that took a while to chase down the part numbers. I wanted to be able to find the correct connector to mate with the unit to preserve its full level of functionality. Any suggestions of where to look for probes and alligator clips? Now to make up some cables. Alternatively, are cables available from Solartron or another supplier? Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of m k Sent: Thursday, August 04, 2011 1:11 AM To: volt-nuts@febo.com Subject: Re: [volt-nuts] Solartron 7081 Test Leads Hi Joe, You will need tellurium copper banana plugs to control the thermal emf's at the meter end and consider the materials that the alligator clips are made of for unintended thermocouples with the copper leads, teflon is a good insulation to have. Mike > From: jltran@att.net > To: volt-nuts@febo.com > Date: Wed, 3 Aug 2011 17:25:07 -0500 > Subject: [volt-nuts] Solartron 7081 Test Leads > > I hope to have a 7081 in the next few weeks that is minus the test leads and > calibration key. I have been able to chase down the connector for the front > panel test leads and two are on the way. However, in fabricating some test > leads, I was wondering if there is anything special about the wires/cable > used to make up the test leads. Silver wire? Teflon insulation? Etc.? > > > > I was planning on using 18 ga. Teflon insulated stranded wire used in > aircraft and some standard 'probes'. For the Kelvin leads, I was thinking > about using alligator clips. Any thoughts, insight, suggestions, previous > experience, etc., would be appreciated. > > > > Thanks in advance. > > > > Joe > > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
JL
J. L. Trantham
Thu, Aug 4, 2011 12:03 PM

Fred,

I'll post an update when I get the unit and the connectors.  If you want the
part numbers of what I ordered, it is:

Part Number                Item
S 104 A053-130+            Connector with solder pins
E3 104.3/6.7+B              Shielded Cable Clamp (non original)
E3 104.6/6.7                Unshielded Cable Clamp (Original)

You can find the Fischer catalog at:

www.fischerconnectors.com/pdf/catalogue/en/Core_Series_Catalogue_v1.pdf

Is your cable from Solartron?  If so, can you send me a close up picture of
the connector end and the 'probe' end of the cable?  Would be nice to see
what it looked like from the factory.

Thanks,

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of Fred Schneider
Sent: Thursday, August 04, 2011 1:16 AM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Solartron 7081 Test Leads

Where did you find the connector, I have a 7060 and It would be nice to have
a second cable.

Fred PA4TIM

Op 4 aug. 2011 om 08:11 heeft m k m1k3k1@hotmail.com het volgende
geschreven:

Hi Joe,

You will need tellurium copper banana plugs to control the thermal emf's

at the meter end and consider the materials that the alligator clips are
made of for unintended thermocouples with the copper leads, teflon is a good
insulation to have.

Mike

From: jltran@att.net
To: volt-nuts@febo.com
Date: Wed, 3 Aug 2011 17:25:07 -0500
Subject: [volt-nuts] Solartron 7081 Test Leads

I hope to have a 7081 in the next few weeks that is minus the test leads

and

calibration key.  I have been able to chase down the connector for the

front

panel test leads and two are on the way.  However, in fabricating some

test

leads, I was wondering if there is anything special about the wires/cable
used to make up the test leads.  Silver wire?  Teflon insulation?  Etc.?

I was planning on using 18 ga. Teflon insulated stranded wire used in
aircraft and some standard 'probes'.  For the Kelvin leads, I was

thinking

about using alligator clips.  Any thoughts, insight, suggestions,

previous

experience, etc., would be appreciated.

Thanks in advance.

Joe


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Fred, I'll post an update when I get the unit and the connectors. If you want the part numbers of what I ordered, it is: Part Number Item S 104 A053-130+ Connector with solder pins E3 104.3/6.7+B Shielded Cable Clamp (non original) E3 104.6/6.7 Unshielded Cable Clamp (Original) You can find the Fischer catalog at: www.fischerconnectors.com/pdf/catalogue/en/Core_Series_Catalogue_v1.pdf Is your cable from Solartron? If so, can you send me a close up picture of the connector end and the 'probe' end of the cable? Would be nice to see what it looked like from the factory. Thanks, Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of Fred Schneider Sent: Thursday, August 04, 2011 1:16 AM To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] Solartron 7081 Test Leads Where did you find the connector, I have a 7060 and It would be nice to have a second cable. Fred PA4TIM Op 4 aug. 2011 om 08:11 heeft m k <m1k3k1@hotmail.com> het volgende geschreven: > > Hi Joe, > > You will need tellurium copper banana plugs to control the thermal emf's at the meter end and consider the materials that the alligator clips are made of for unintended thermocouples with the copper leads, teflon is a good insulation to have. > > Mike > >> From: jltran@att.net >> To: volt-nuts@febo.com >> Date: Wed, 3 Aug 2011 17:25:07 -0500 >> Subject: [volt-nuts] Solartron 7081 Test Leads >> >> I hope to have a 7081 in the next few weeks that is minus the test leads and >> calibration key. I have been able to chase down the connector for the front >> panel test leads and two are on the way. However, in fabricating some test >> leads, I was wondering if there is anything special about the wires/cable >> used to make up the test leads. Silver wire? Teflon insulation? Etc.? >> >> >> >> I was planning on using 18 ga. Teflon insulated stranded wire used in >> aircraft and some standard 'probes'. For the Kelvin leads, I was thinking >> about using alligator clips. Any thoughts, insight, suggestions, previous >> experience, etc., would be appreciated. >> >> >> >> Thanks in advance. >> >> >> >> Joe >> >> >> >> _______________________________________________ >> volt-nuts mailing list -- volt-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts >> and follow the instructions there. > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
RA
Robert Atkinson
Thu, Aug 4, 2011 5:13 PM

Hi,
NPL recommend Van Damme copper cable. < http://www.vdctrading.com/ >.They make a nice quad core. It's a professional (rather than audiofool) audio cable. PTFE cable has no advantage. It's plated and stiff.
Pomona make TeCu binding posts and plugs tHe binding post is 3770 and cost about £6 each.

Robert G8rPI.

--- On Thu, 4/8/11, J. L. Trantham jltran@att.net wrote:

From: J. L. Trantham jltran@att.net
Subject: Re: [volt-nuts] Solartron 7081 Test Leads
To: "'Discussion of precise voltage measurement'" volt-nuts@febo.com
Date: Thursday, 4 August, 2011, 12:50

Mike,

I am new to the arena of high accuracy measurements but the plan is for this
to be the 'lab standard'.

There are no banana sockets on the front of the 7081.  It is a Fischer 5 pin
connector that took a while to chase down the part numbers.  I wanted to be
able to find the correct connector to mate with the unit to preserve its
full level of functionality.

Any suggestions of where to look for probes and alligator clips?

Now to make up some cables.  Alternatively, are cables available from
Solartron or another supplier?

Joe

-----Original Message-----
From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On
Behalf Of m k
Sent: Thursday, August 04, 2011 1:11 AM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Solartron 7081 Test Leads

Hi Joe,

You will need tellurium copper banana plugs to control the thermal emf's at
the meter end and consider the materials that the alligator clips are made
of for unintended thermocouples with the copper leads, teflon is a good
insulation to have.

Mike

From: jltran@att.net
To: volt-nuts@febo.com
Date: Wed, 3 Aug 2011 17:25:07 -0500
Subject: [volt-nuts] Solartron 7081 Test Leads

I hope to have a 7081 in the next few weeks that is minus the test leads

and

calibration key.  I have been able to chase down the connector for the

front

panel test leads and two are on the way.  However, in fabricating some

test

leads, I was wondering if there is anything special about the wires/cable
used to make up the test leads.  Silver wire?  Teflon insulation?  Etc.?

 

I was planning on using 18 ga. Teflon insulated stranded wire used in
aircraft and some standard 'probes'.  For the Kelvin leads, I was thinking
about using alligator clips.  Any thoughts, insight, suggestions, previous
experience, etc., would be appreciated.

 

Thanks in advance.

 

Joe

 


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to

and follow the instructions there.

                         


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hi, NPL recommend Van Damme copper cable. < http://www.vdctrading.com/ >.They make a nice quad core. It's a professional (rather than audiofool) audio cable. PTFE cable has no advantage. It's plated and stiff. Pomona make TeCu binding posts and plugs tHe binding post is 3770 and cost about £6 each. Robert G8rPI. --- On Thu, 4/8/11, J. L. Trantham <jltran@att.net> wrote: From: J. L. Trantham <jltran@att.net> Subject: Re: [volt-nuts] Solartron 7081 Test Leads To: "'Discussion of precise voltage measurement'" <volt-nuts@febo.com> Date: Thursday, 4 August, 2011, 12:50 Mike, I am new to the arena of high accuracy measurements but the plan is for this to be the 'lab standard'. There are no banana sockets on the front of the 7081.  It is a Fischer 5 pin connector that took a while to chase down the part numbers.  I wanted to be able to find the correct connector to mate with the unit to preserve its full level of functionality. Any suggestions of where to look for probes and alligator clips? Now to make up some cables.  Alternatively, are cables available from Solartron or another supplier? Joe -----Original Message----- From: volt-nuts-bounces@febo.com [mailto:volt-nuts-bounces@febo.com] On Behalf Of m k Sent: Thursday, August 04, 2011 1:11 AM To: volt-nuts@febo.com Subject: Re: [volt-nuts] Solartron 7081 Test Leads Hi Joe, You will need tellurium copper banana plugs to control the thermal emf's at the meter end and consider the materials that the alligator clips are made of for unintended thermocouples with the copper leads, teflon is a good insulation to have. Mike > From: jltran@att.net > To: volt-nuts@febo.com > Date: Wed, 3 Aug 2011 17:25:07 -0500 > Subject: [volt-nuts] Solartron 7081 Test Leads > > I hope to have a 7081 in the next few weeks that is minus the test leads and > calibration key.  I have been able to chase down the connector for the front > panel test leads and two are on the way.  However, in fabricating some test > leads, I was wondering if there is anything special about the wires/cable > used to make up the test leads.  Silver wire?  Teflon insulation?  Etc.? > >  > > I was planning on using 18 ga. Teflon insulated stranded wire used in > aircraft and some standard 'probes'.  For the Kelvin leads, I was thinking > about using alligator clips.  Any thoughts, insight, suggestions, previous > experience, etc., would be appreciated. > >  > > Thanks in advance. > >  > > Joe > >  > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.                           _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.