HP5370 power supply measurements

CS
Charles Steinmetz
Fri, Jul 15, 2016 2:44 PM

David wrote:

An external fan blowing on the heatsink keeps my
5370s nice and cool, for about $10 each.

So more noise.

You seem to have missed the next sentence, which said they are
"extremely quiet."  So quiet that I could reduce the noise of the
internal fan by 20dB and still not be able to tell whether the external
fan was running without looking at it. [1]  So no, as a practical matter
there is no additional noise.

Surely one of the main  advantages of the SMPS route is the higher
efficiently means less cooling air is required.  That means less noise.

Two different things.  The internal fan primarily cools the inside of
the counter, while the power supply heat is developed on the heatsink,
which is mainly cooled by convection outside the enclosure.  (Yes, the
heatsink has a couple of little holes in it where air from inside the
counter blows out -- but that is completely ineffectual at reducing the
heatsink temperature, and I doubt that's what the designers intended.
More likely, they were trying to minimize the amount of heat from the
heatsink that gets inside the box, by providing a path for air to flow
from the inside, past the heatsink, to the outside -- and at that, the
holes seem effective.)  So, if you reduce the internal fan airflow, the
internal circuitry will run hotter even though the power supply pass
devices won't (much).  Further, if you replace the heatsink-mounted
regulators with switchers that are located inside the box, you may very
well put MORE heat load on the internal fan even though the switchers
are much more efficient (because what loss there is would be dissipated
inside the box, rather than outside).

Best regards,

Charles

[1]  I actually tried an internal fan that reduced the noise by 22dB,
but it didn't have sufficient airflow to keep the interior properly cool.

David wrote: >> An external fan blowing on the heatsink keeps my >> 5370s nice and cool, for about $10 each. > > So more noise. You seem to have missed the next sentence, which said they are "extremely quiet." So quiet that I could reduce the noise of the internal fan by 20dB and still not be able to tell whether the external fan was running without looking at it. [1] So no, as a practical matter there is no additional noise. > Surely one of the main advantages of the SMPS route is the higher > efficiently means less cooling air is required. That means less noise. Two different things. The internal fan primarily cools the inside of the counter, while the power supply heat is developed on the heatsink, which is mainly cooled by convection outside the enclosure. (Yes, the heatsink has a couple of little holes in it where air from inside the counter blows out -- but that is completely ineffectual at reducing the heatsink temperature, and I doubt that's what the designers intended. More likely, they were trying to minimize the amount of heat from the heatsink that gets inside the box, by providing a path for air to flow from the inside, past the heatsink, to the outside -- and at that, the holes seem effective.) So, if you reduce the internal fan airflow, the internal circuitry will run hotter even though the power supply pass devices won't (much). Further, if you replace the heatsink-mounted regulators with switchers that are located inside the box, you may very well put MORE heat load on the internal fan even though the switchers are much more efficient (because what loss there is would be dissipated inside the box, rather than outside). Best regards, Charles [1] I actually tried an internal fan that reduced the noise by 22dB, but it didn't have sufficient airflow to keep the interior properly cool.
PK
Poul-Henning Kamp
Fri, Jul 15, 2016 5:55 PM

David Kirkby scripsit:

I often here of people replacing fans with quiter ones, but I suspect that
all they really do is reduce the airflow.

Not necessarily, Dave. The Austrian company, Noctua, for one, makes
extremely quiet fans with excellent airflow.

... at zero pressure differential, which is easy to do (Think: ceiling fan).

The hard part is to make fans silently move air when they also have
to build up pressure to get the air to move places.

This is where the gap between the wings and the frame must be minimized,
or the pressure will leak out through it.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

-------- In message <CANy2iXq6ONvriDoFgNFkEBQjDKntP7t8KuE7BOUpxJWLCuxLsw@mail.gmail.com> , "William H. Fite" writes: >David Kirkby scripsit: >> >> I often here of people replacing fans with quiter ones, but I suspect that >> all they really do is reduce the airflow. > >Not necessarily, Dave. The Austrian company, Noctua, for one, makes >extremely quiet fans with excellent airflow. ... at zero pressure differential, which is easy to do (Think: ceiling fan). The hard part is to make fans silently move air when they also have to build up pressure to get the air to move places. This is where the gap between the wings and the frame must be minimized, or the pressure will leak out through it. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.
OE
Orin Eman
Fri, Jul 15, 2016 6:40 PM

On Fri, Jul 15, 2016 at 10:55 AM, Poul-Henning Kamp phk@phk.freebsd.dk
wrote:


In message <
CANy2iXq6ONvriDoFgNFkEBQjDKntP7t8KuE7BOUpxJWLCuxLsw@mail.gmail.com>
, "William H. Fite" writes:

David Kirkby scripsit:

I often here of people replacing fans with quiter ones, but I suspect

that

all they really do is reduce the airflow.

Not necessarily, Dave. The Austrian company, Noctua, for one, makes
extremely quiet fans with excellent airflow.

... at zero pressure differential, which is easy to do (Think: ceiling
fan).

Right.

You have to look at the curves on the data sheet that shows air flow vs.
static pressure (and be careful about the static pressure scale).  I found
that a 'quiet' fan would often be flowing one tenth as much air as the
original fan at the static pressure at which the original fan was rated.

In a given instrument, you may get away with the quieter fan, but how would
you tell other than putting a thermometer inside and making a before/after
comparison?

Orin.

On Fri, Jul 15, 2016 at 10:55 AM, Poul-Henning Kamp <phk@phk.freebsd.dk> wrote: > -------- > In message < > CANy2iXq6ONvriDoFgNFkEBQjDKntP7t8KuE7BOUpxJWLCuxLsw@mail.gmail.com> > , "William H. Fite" writes: > >David Kirkby scripsit: > >> > >> I often here of people replacing fans with quiter ones, but I suspect > that > >> all they really do is reduce the airflow. > > > >Not necessarily, Dave. The Austrian company, Noctua, for one, makes > >extremely quiet fans with excellent airflow. > > ... at zero pressure differential, which is easy to do (Think: ceiling > fan). > > Right. You have to look at the curves on the data sheet that shows air flow vs. static pressure (and be careful about the static pressure scale). I found that a 'quiet' fan would often be flowing one tenth as much air as the original fan at the static pressure at which the original fan was rated. In a given instrument, you may get away with the quieter fan, but how would you tell other than putting a thermometer inside and making a before/after comparison? Orin.
LV
Lester Veenstra
Fri, Jul 15, 2016 6:46 PM

Anybody want GPS Antenna Micropulse Z1001?
$20 each plus shipping

Lester B Veenstra  MØYCM K1YCM W8YCM
lester@veenstras.com

Physical and US Postal Addresses
5 Shrine Club Drive (Physical)
452 Stable Ln (RFD USPS Mail)
Keyser WV 26726
GPS: 39.336826 N  78.982287 W (Google)
GPS: 39.33682 N  78.9823741 W (GPSDO)

Telephones:
Home:                    +1-304-289-6057
US cell                    +1-304-790-9192
UK cell                    +44-(0)7849-248-749
Guam Cell:              +1-671-929-8141
Jamaica:                +1-876-456-8898

Anybody want GPS Antenna Micropulse Z1001? $20 each plus shipping Lester B Veenstra MØYCM K1YCM W8YCM lester@veenstras.com Physical and US Postal Addresses 5 Shrine Club Drive (Physical) 452 Stable Ln (RFD USPS Mail) Keyser WV 26726 GPS: 39.336826 N 78.982287 W (Google) GPS: 39.33682 N 78.9823741 W (GPSDO) Telephones: Home: +1-304-289-6057 US cell +1-304-790-9192 UK cell +44-(0)7849-248-749 Guam Cell: +1-671-929-8141 Jamaica: +1-876-456-8898
WH
William H. Fite
Fri, Jul 15, 2016 7:22 PM

The fact remains that there are some very quiet equipment fans besides the
little muffins that sound like drones lifting off.

On Friday, July 15, 2016, Poul-Henning Kamp phk@phk.freebsd.dk wrote:

David Kirkby scripsit:

I often here of people replacing fans with quiter ones, but I suspect

that

all they really do is reduce the airflow.

Not necessarily, Dave. The Austrian company, Noctua, for one, makes
extremely quiet fans with excellent airflow.

... at zero pressure differential, which is easy to do (Think: ceiling
fan).

The hard part is to make fans silently move air when they also have
to build up pressure to get the air to move places.

This is where the gap between the wings and the frame must be minimized,
or the pressure will leak out through it.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

--
Yes, Black lives do matter.

The fact remains that there are some very quiet equipment fans besides the little muffins that sound like drones lifting off. On Friday, July 15, 2016, Poul-Henning Kamp <phk@phk.freebsd.dk> wrote: > -------- > In message < > CANy2iXq6ONvriDoFgNFkEBQjDKntP7t8KuE7BOUpxJWLCuxLsw@mail.gmail.com > <javascript:;>> > , "William H. Fite" writes: > >David Kirkby scripsit: > >> > >> I often here of people replacing fans with quiter ones, but I suspect > that > >> all they really do is reduce the airflow. > > > >Not necessarily, Dave. The Austrian company, Noctua, for one, makes > >extremely quiet fans with excellent airflow. > > ... at zero pressure differential, which is easy to do (Think: ceiling > fan). > > The hard part is to make fans silently move air when they also have > to build up pressure to get the air to move places. > > This is where the gap between the wings and the frame must be minimized, > or the pressure will leak out through it. > > -- > Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 > phk@FreeBSD.ORG | TCP/IP since RFC 956 > FreeBSD committer | BSD since 4.3-tahoe > Never attribute to malice what can adequately be explained by incompetence. > -- Yes, Black lives do matter.
WH
William H. Fite
Fri, Jul 15, 2016 7:52 PM

That is, in fact, precisely how you do it.

On Friday, July 15, 2016, Orin Eman orin.eman@gmail.com wrote:

On Fri, Jul 15, 2016 at 10:55 AM, Poul-Henning Kamp <phk@phk.freebsd.dk
javascript:;>
wrote:

, "William H. Fite" writes:

David Kirkby scripsit:

I often here of people replacing fans with quiter ones, but I suspect

that

all they really do is reduce the airflow.

Not necessarily, Dave. The Austrian company, Noctua, for one, makes
extremely quiet fans with excellent airflow.

... at zero pressure differential, which is easy to do (Think: ceiling
fan).

Right.

You have to look at the curves on the data sheet that shows air flow vs.
static pressure (and be careful about the static pressure scale).  I found
that a 'quiet' fan would often be flowing one tenth as much air as the
original fan at the static pressure at which the original fan was rated.

In a given instrument, you may get away with the quieter fan, but how would
you tell other than putting a thermometer inside and making a before/after
comparison?

Orin.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
Yes, Black lives do matter.

That is, in fact, precisely how you do it. On Friday, July 15, 2016, Orin Eman <orin.eman@gmail.com> wrote: > On Fri, Jul 15, 2016 at 10:55 AM, Poul-Henning Kamp <phk@phk.freebsd.dk > <javascript:;>> > wrote: > > > -------- > > In message < > > CANy2iXq6ONvriDoFgNFkEBQjDKntP7t8KuE7BOUpxJWLCuxLsw@mail.gmail.com > <javascript:;>> > > , "William H. Fite" writes: > > >David Kirkby scripsit: > > >> > > >> I often here of people replacing fans with quiter ones, but I suspect > > that > > >> all they really do is reduce the airflow. > > > > > >Not necessarily, Dave. The Austrian company, Noctua, for one, makes > > >extremely quiet fans with excellent airflow. > > > > ... at zero pressure differential, which is easy to do (Think: ceiling > > fan). > > > > > > Right. > > You have to look at the curves on the data sheet that shows air flow vs. > static pressure (and be careful about the static pressure scale). I found > that a 'quiet' fan would often be flowing one tenth as much air as the > original fan at the static pressure at which the original fan was rated. > > In a given instrument, you may get away with the quieter fan, but how would > you tell other than putting a thermometer inside and making a before/after > comparison? > > Orin. > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com <javascript:;> > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > -- Yes, Black lives do matter.
CA
Clay Autery
Fri, Jul 15, 2016 8:17 PM

Actually, the lion's share of noise heard from a fan is NOT the moving
air, but the turbulence over the fan blades... This is a VERY large
topic that I don't want to bore the list with.


Clay Autery, KY5G
MONTAC Enterprises
(318) 518-1389

On 7/15/2016 8:17 AM, Dr. David Kirkby (Kirkby Microwave Ltd) wrote:

I often here of people replacing fans with quiter ones, but I suspect that
all they really do is reduce the airflow. I believe that most of the noise
one hears is the movement of the air.  From what I have read, sleve bearing
fans make a bit less noise than ball bearings fans.  However,  although the
MTBF of both types is similar at 20 degree C, the MTBF of sleve bearing
fans decreases quite a bit with only a modest increase in temperature.

Dave.

Actually, the lion's share of noise heard from a fan is NOT the moving air, but the turbulence over the fan blades... This is a VERY large topic that I don't want to bore the list with. ______________________ Clay Autery, KY5G MONTAC Enterprises (318) 518-1389 On 7/15/2016 8:17 AM, Dr. David Kirkby (Kirkby Microwave Ltd) wrote: > > I often here of people replacing fans with quiter ones, but I suspect that > all they really do is reduce the airflow. I believe that most of the noise > one hears is the movement of the air. From what I have read, sleve bearing > fans make a bit less noise than ball bearings fans. However, although the > MTBF of both types is similar at 20 degree C, the MTBF of sleve bearing > fans decreases quite a bit with only a modest increase in temperature. > > Dave.
PK
Poul-Henning Kamp
Fri, Jul 15, 2016 8:22 PM

The fact remains that there are some very quiet equipment fans besides the
little muffins that sound like drones lifting off.

Yes, there are absolutely big differences between fans of same
physical footprint.

But most of the "so quiet we had to break the laws of nature" fans
aimed at the PC market don't do anything like that, and it would
be a big mistake to use such a fan to replace the HP5370 fan, where
a significant pressure is required.

For attaching to the cooling fins to get some air moving, they're
fine however.

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

-------- >The fact remains that there are some very quiet equipment fans besides the >little muffins that sound like drones lifting off. Yes, there are absolutely big differences between fans of same physical footprint. But most of the "so quiet we had to break the laws of nature" fans aimed at the PC market don't do anything like that, and it would be a big mistake to use such a fan to replace the HP5370 fan, where a significant pressure is required. For attaching to the cooling fins to get some air moving, they're fine however. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.
TP
Tommy Phone
Fri, Jul 15, 2016 8:25 PM

Some thirty years ago fans for automotive radiator cooling were designed with 7 unevenly spaced blades to reduce the siren effect yet yield comparable air flow. Careful layout and blade sweep back along with an annular ring made it entirely feasible to have a statically and dynamically balanced fan as it came out of the mold. I always wondered why the folks who make these little fans can't figure out how to do that. Maybe getting comparable CFM from a much smaller fan violates some Reynolds number requirement for turbulence control.

From Tom Holmes, N8ZM

On Jul 15, 2016, at 3:52 PM, William H. Fite omniryx@gmail.com wrote:

That is, in fact, precisely how you do it.

On Friday, July 15, 2016, Orin Eman orin.eman@gmail.com wrote:

On Fri, Jul 15, 2016 at 10:55 AM, Poul-Henning Kamp <phk@phk.freebsd.dk
javascript:;>
wrote:

, "William H. Fite" writes:

David Kirkby scripsit:

I often here of people replacing fans with quiter ones, but I suspect

that

all they really do is reduce the airflow.

Not necessarily, Dave. The Austrian company, Noctua, for one, makes
extremely quiet fans with excellent airflow.

... at zero pressure differential, which is easy to do (Think: ceiling
fan).

Right.

You have to look at the curves on the data sheet that shows air flow vs.
static pressure (and be careful about the static pressure scale).  I found
that a 'quiet' fan would often be flowing one tenth as much air as the
original fan at the static pressure at which the original fan was rated.

In a given instrument, you may get away with the quieter fan, but how would
you tell other than putting a thermometer inside and making a before/after
comparison?

Orin.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
Yes, Black lives do matter.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Some thirty years ago fans for automotive radiator cooling were designed with 7 unevenly spaced blades to reduce the siren effect yet yield comparable air flow. Careful layout and blade sweep back along with an annular ring made it entirely feasible to have a statically and dynamically balanced fan as it came out of the mold. I always wondered why the folks who make these little fans can't figure out how to do that. Maybe getting comparable CFM from a much smaller fan violates some Reynolds number requirement for turbulence control. From Tom Holmes, N8ZM > On Jul 15, 2016, at 3:52 PM, William H. Fite <omniryx@gmail.com> wrote: > > That is, in fact, precisely how you do it. > > >> On Friday, July 15, 2016, Orin Eman <orin.eman@gmail.com> wrote: >> >> On Fri, Jul 15, 2016 at 10:55 AM, Poul-Henning Kamp <phk@phk.freebsd.dk >> <javascript:;>> >> wrote: >> >>> -------- >>> In message < >>> CANy2iXq6ONvriDoFgNFkEBQjDKntP7t8KuE7BOUpxJWLCuxLsw@mail.gmail.com >> <javascript:;>> >>> , "William H. Fite" writes: >>>> David Kirkby scripsit: >>>>> >>>>> I often here of people replacing fans with quiter ones, but I suspect >>> that >>>>> all they really do is reduce the airflow. >>>> >>>> Not necessarily, Dave. The Austrian company, Noctua, for one, makes >>>> extremely quiet fans with excellent airflow. >>> >>> ... at zero pressure differential, which is easy to do (Think: ceiling >>> fan). >>> >>> >> >> Right. >> >> You have to look at the curves on the data sheet that shows air flow vs. >> static pressure (and be careful about the static pressure scale). I found >> that a 'quiet' fan would often be flowing one tenth as much air as the >> original fan at the static pressure at which the original fan was rated. >> >> In a given instrument, you may get away with the quieter fan, but how would >> you tell other than putting a thermometer inside and making a before/after >> comparison? >> >> Orin. >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >> To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > > > -- > Yes, Black lives do matter. > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BC
Bob Camp
Fri, Jul 15, 2016 8:59 PM

Hi

Back when these pieces of gear were being made “new”, HP Spokane had a
“fan silencer machine”. They used a strobe and a microphone to identify imbalanced
blades and then took chunks out of them. They claimed it made the fans significantly
less noisy….Watching it in action on a noisy factory floor … not so clear.

Bob

On Jul 15, 2016, at 4:25 PM, Tommy Phone tholmes@woh.rr.com wrote:

Some thirty years ago fans for automotive radiator cooling were designed with 7 unevenly spaced blades to reduce the siren effect yet yield comparable air flow. Careful layout and blade sweep back along with an annular ring made it entirely feasible to have a statically and dynamically balanced fan as it came out of the mold. I always wondered why the folks who make these little fans can't figure out how to do that. Maybe getting comparable CFM from a much smaller fan violates some Reynolds number requirement for turbulence control.

From Tom Holmes, N8ZM

On Jul 15, 2016, at 3:52 PM, William H. Fite omniryx@gmail.com wrote:

That is, in fact, precisely how you do it.

On Friday, July 15, 2016, Orin Eman orin.eman@gmail.com wrote:

On Fri, Jul 15, 2016 at 10:55 AM, Poul-Henning Kamp <phk@phk.freebsd.dk
javascript:;>
wrote:

, "William H. Fite" writes:

David Kirkby scripsit:

I often here of people replacing fans with quiter ones, but I suspect

that

all they really do is reduce the airflow.

Not necessarily, Dave. The Austrian company, Noctua, for one, makes
extremely quiet fans with excellent airflow.

... at zero pressure differential, which is easy to do (Think: ceiling
fan).

Right.

You have to look at the curves on the data sheet that shows air flow vs.
static pressure (and be careful about the static pressure scale).  I found
that a 'quiet' fan would often be flowing one tenth as much air as the
original fan at the static pressure at which the original fan was rated.

In a given instrument, you may get away with the quieter fan, but how would
you tell other than putting a thermometer inside and making a before/after
comparison?

Orin.


time-nuts mailing list -- time-nuts@febo.com javascript:;
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

--
Yes, Black lives do matter.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Back when these pieces of gear were being made “new”, HP Spokane had a “fan silencer machine”. They used a strobe and a microphone to identify imbalanced blades and then took chunks out of them. They *claimed* it made the fans significantly less noisy….Watching it in action on a noisy factory floor … not so clear. Bob > On Jul 15, 2016, at 4:25 PM, Tommy Phone <tholmes@woh.rr.com> wrote: > > Some thirty years ago fans for automotive radiator cooling were designed with 7 unevenly spaced blades to reduce the siren effect yet yield comparable air flow. Careful layout and blade sweep back along with an annular ring made it entirely feasible to have a statically and dynamically balanced fan as it came out of the mold. I always wondered why the folks who make these little fans can't figure out how to do that. Maybe getting comparable CFM from a much smaller fan violates some Reynolds number requirement for turbulence control. > > From Tom Holmes, N8ZM > >> On Jul 15, 2016, at 3:52 PM, William H. Fite <omniryx@gmail.com> wrote: >> >> That is, in fact, precisely how you do it. >> >> >>> On Friday, July 15, 2016, Orin Eman <orin.eman@gmail.com> wrote: >>> >>> On Fri, Jul 15, 2016 at 10:55 AM, Poul-Henning Kamp <phk@phk.freebsd.dk >>> <javascript:;>> >>> wrote: >>> >>>> -------- >>>> In message < >>>> CANy2iXq6ONvriDoFgNFkEBQjDKntP7t8KuE7BOUpxJWLCuxLsw@mail.gmail.com >>> <javascript:;>> >>>> , "William H. Fite" writes: >>>>> David Kirkby scripsit: >>>>>> >>>>>> I often here of people replacing fans with quiter ones, but I suspect >>>> that >>>>>> all they really do is reduce the airflow. >>>>> >>>>> Not necessarily, Dave. The Austrian company, Noctua, for one, makes >>>>> extremely quiet fans with excellent airflow. >>>> >>>> ... at zero pressure differential, which is easy to do (Think: ceiling >>>> fan). >>>> >>>> >>> >>> Right. >>> >>> You have to look at the curves on the data sheet that shows air flow vs. >>> static pressure (and be careful about the static pressure scale). I found >>> that a 'quiet' fan would often be flowing one tenth as much air as the >>> original fan at the static pressure at which the original fan was rated. >>> >>> In a given instrument, you may get away with the quieter fan, but how would >>> you tell other than putting a thermometer inside and making a before/after >>> comparison? >>> >>> Orin. >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com <javascript:;> >>> To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >> >> >> -- >> Yes, Black lives do matter. >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.