5370B Question / help needed

TA
Thomas Allgeier
Sat, Jul 1, 2017 8:12 PM

I got hold of a 5370B cheaply but it turns out it may have a fault:

Going through the motions of measuring TI with it's own 10MHz ref as input
(as described in the manual) I don't get 100 ns, but around 15 ns. So this
is with the switch set to START COM. Oddly enough I get the same 15 ns with
the switch set to SEP, and going through a 6 ft cable between start and
stop.

Now I can "tune" the thing to display 100 ns by changing the trigger
levels. Start at 0.6 V and stop at -0.1 V will achieve this. Once again
this does not change by going from COM to SEP i,e. by adding the 6 ft delay.

One other observation, probably related: After powerup it took several
vigorous switch activations of the COM/SEP switch before the thing swung
into action at all - i.e. the first reaction with COM setting was an Err
02, and with SEP I got the 15 ns as described above. So that switch may be
a bit dodgy.

Measuring in frequency mode gives the 10 MHz quite accurately, less than
1mHz off with its own ref as stop input. Feeding the 10 MHz in from my
GPSDO also ties up OK, it is around 40mHz off with perhaps 5-10mHz jitter.

Has anybody come across anything similar? I couldn't find much in that vein
in the archive but may have overlooked something. The one thing I was
worried about at first was the heat sinks at the rear getting really hot
but it seems I'm not the only one with that observation.

To some extent it looks like the input levels for triggering are off, but
then the fact that the 6 ft cable don't show up as extra delay may mean
there is something else wrong.

Looks like some "fun with Bill and Dave" lies ahead if somebody can point
me in the right direction.

Thanks in advance,
Thomas.

I got hold of a 5370B cheaply but it turns out it may have a fault: Going through the motions of measuring TI with it's own 10MHz ref as input (as described in the manual) I don't get 100 ns, but around 15 ns. So this is with the switch set to START COM. Oddly enough I get the same 15 ns with the switch set to SEP, and going through a 6 ft cable between start and stop. Now I can "tune" the thing to display 100 ns by changing the trigger levels. Start at 0.6 V and stop at -0.1 V will achieve this. Once again this does not change by going from COM to SEP i,e. by adding the 6 ft delay. One other observation, probably related: After powerup it took several vigorous switch activations of the COM/SEP switch before the thing swung into action at all - i.e. the first reaction with COM setting was an Err 02, and with SEP I got the 15 ns as described above. So that switch may be a bit dodgy. Measuring in frequency mode gives the 10 MHz quite accurately, less than 1mHz off with its own ref as stop input. Feeding the 10 MHz in from my GPSDO also ties up OK, it is around 40mHz off with perhaps 5-10mHz jitter. Has anybody come across anything similar? I couldn't find much in that vein in the archive but may have overlooked something. The one thing I was worried about at first was the heat sinks at the rear getting really hot but it seems I'm not the only one with that observation. To some extent it looks like the input levels for triggering are off, but then the fact that the 6 ft cable don't show up as extra delay may mean there is something else wrong. Looks like some "fun with Bill and Dave" lies ahead if somebody can point me in the right direction. Thanks in advance, Thomas.
PK
Poul-Henning Kamp
Sat, Jul 1, 2017 8:38 PM

Going through the motions of measuring TI with it's own 10MHz ref as input
(as described in the manual) I don't get 100 ns, but around 15 ns. So this
is with the switch set to START COM. Oddly enough I get the same 15 ns with
the switch set to SEP, and going through a 6 ft cable between start and
stop.

Have you checked the trigger polarities ?

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

-------- In message <CAOT1cQJ4K8uyNvFp8AjoLGjd57Pm31LnZT_jRo44J4Zhesvgzg@mail.gmail.com>, Thomas Allgeier writes: >Going through the motions of measuring TI with it's own 10MHz ref as input >(as described in the manual) I don't get 100 ns, but around 15 ns. So this >is with the switch set to START COM. Oddly enough I get the same 15 ns with >the switch set to SEP, and going through a 6 ft cable between start and >stop. Have you checked the trigger polarities ? -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 phk@FreeBSD.ORG | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence.
TA
Thomas Allgeier
Sat, Jul 1, 2017 10:16 PM

Yes, they are both set to "positive".  Leaving the start at "pos" and
switching the stop to "neg" gives 66 ns. One more observation, which
probably does make sense: If I set the ARming to +/-TI and toggle the
Period Complement the "sum" is always very close to 100ns: 15 is
complemented by -85 and 66 by -34. All this is for Trigger Levels at Preset.

When I "tune" the levels to get 100 ns at +TI and then switch to +/-TI I
get around 20 ps, and the Complement button has no effect on this.

On Sat, Jul 1, 2017 at 9:38 PM, Poul-Henning Kamp phk@phk.freebsd.dk
wrote:


In message <CAOT1cQJ4K8uyNvFp8AjoLGjd57Pm31LnZT_jRo44J4Zhesvgzg@mail.
gmail.com>, Thomas Allgeier writes:

Going through the motions of measuring TI with it's own 10MHz ref as input
(as described in the manual) I don't get 100 ns, but around 15 ns. So this
is with the switch set to START COM. Oddly enough I get the same 15 ns

with

the switch set to SEP, and going through a 6 ft cable between start and
stop.

Have you checked the trigger polarities ?

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

Yes, they are both set to "positive". Leaving the start at "pos" and switching the stop to "neg" gives 66 ns. One more observation, which probably does make sense: If I set the ARming to +/-TI and toggle the Period Complement the "sum" is always very close to 100ns: 15 is complemented by -85 and 66 by -34. All this is for Trigger Levels at Preset. When I "tune" the levels to get 100 ns at +TI and then switch to +/-TI I get around 20 ps, and the Complement button has no effect on this. On Sat, Jul 1, 2017 at 9:38 PM, Poul-Henning Kamp <phk@phk.freebsd.dk> wrote: > -------- > In message <CAOT1cQJ4K8uyNvFp8AjoLGjd57Pm31LnZT_jRo44J4Zhesvgzg@mail. > gmail.com>, Thomas Allgeier writes: > > >Going through the motions of measuring TI with it's own 10MHz ref as input > >(as described in the manual) I don't get 100 ns, but around 15 ns. So this > >is with the switch set to START COM. Oddly enough I get the same 15 ns > with > >the switch set to SEP, and going through a 6 ft cable between start and > >stop. > > Have you checked the trigger polarities ? > > > -- > Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 > phk@FreeBSD.ORG | TCP/IP since RFC 956 > FreeBSD committer | BSD since 4.3-tahoe > Never attribute to malice what can adequately be explained by incompetence. >
MD
Magnus Danielson
Sun, Jul 2, 2017 8:31 AM

Hi,

In the manual there is an adjustment procedure. Have you looked at that?

DC offsets in trigger points do shift the start and stop triggers.

Cheers,
Magnus

On 07/02/2017 12:16 AM, Thomas Allgeier wrote:

Yes, they are both set to "positive".  Leaving the start at "pos" and
switching the stop to "neg" gives 66 ns. One more observation, which
probably does make sense: If I set the ARming to +/-TI and toggle the
Period Complement the "sum" is always very close to 100ns: 15 is
complemented by -85 and 66 by -34. All this is for Trigger Levels at Preset.

When I "tune" the levels to get 100 ns at +TI and then switch to +/-TI I
get around 20 ps, and the Complement button has no effect on this.

On Sat, Jul 1, 2017 at 9:38 PM, Poul-Henning Kamp phk@phk.freebsd.dk
wrote:


In message <CAOT1cQJ4K8uyNvFp8AjoLGjd57Pm31LnZT_jRo44J4Zhesvgzg@mail.
gmail.com>, Thomas Allgeier writes:

Going through the motions of measuring TI with it's own 10MHz ref as input
(as described in the manual) I don't get 100 ns, but around 15 ns. So this
is with the switch set to START COM. Oddly enough I get the same 15 ns

with

the switch set to SEP, and going through a 6 ft cable between start and
stop.

Have you checked the trigger polarities ?

--
Poul-Henning Kamp      | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG        | TCP/IP since RFC 956
FreeBSD committer      | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi, In the manual there is an adjustment procedure. Have you looked at that? DC offsets in trigger points do shift the start and stop triggers. Cheers, Magnus On 07/02/2017 12:16 AM, Thomas Allgeier wrote: > Yes, they are both set to "positive". Leaving the start at "pos" and > switching the stop to "neg" gives 66 ns. One more observation, which > probably does make sense: If I set the ARming to +/-TI and toggle the > Period Complement the "sum" is always very close to 100ns: 15 is > complemented by -85 and 66 by -34. All this is for Trigger Levels at Preset. > > When I "tune" the levels to get 100 ns at +TI and then switch to +/-TI I > get around 20 ps, and the Complement button has no effect on this. > > On Sat, Jul 1, 2017 at 9:38 PM, Poul-Henning Kamp <phk@phk.freebsd.dk> > wrote: > >> -------- >> In message <CAOT1cQJ4K8uyNvFp8AjoLGjd57Pm31LnZT_jRo44J4Zhesvgzg@mail. >> gmail.com>, Thomas Allgeier writes: >> >>> Going through the motions of measuring TI with it's own 10MHz ref as input >>> (as described in the manual) I don't get 100 ns, but around 15 ns. So this >>> is with the switch set to START COM. Oddly enough I get the same 15 ns >> with >>> the switch set to SEP, and going through a 6 ft cable between start and >>> stop. >> >> Have you checked the trigger polarities ? >> >> >> -- >> Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 >> phk@FreeBSD.ORG | TCP/IP since RFC 956 >> FreeBSD committer | BSD since 4.3-tahoe >> Never attribute to malice what can adequately be explained by incompetence. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
GN
Gary Neilson
Mon, Jul 3, 2017 2:13 PM

I have a 5370B from ebay, it was missing a couple of boards and had a
bad interpolator card. Also a bad input hybrid on the stop channel.

My unit acts similar to yours, TI measurement on the 10mhz signal shows
107 ns.

Frequency and Period measurement is spot on.

I think it needs an alignment performed. I will get to that soon.

Good Luck

Gary - K5DSR

-------- Forwarded Message --------
Subject: [time-nuts] 5370B Question / help needed
Date: Sat, 1 Jul 2017 21:12:27 +0100
From: Thomas Allgeier th.allgeier@gmail.com
Reply-To: Discussion of precise time and frequency measurement
time-nuts@febo.com
To: Discussion of precise time and frequency measurement
time-nuts@febo.com

I got hold of a 5370B cheaply but it turns out it may have a fault:

Going through the motions of measuring TI with it's own 10MHz ref as input
(as described in the manual) I don't get 100 ns, but around 15 ns. So this
is with the switch set to START COM. Oddly enough I get the same 15 ns with
the switch set to SEP, and going through a 6 ft cable between start and
stop.

Now I can "tune" the thing to display 100 ns by changing the trigger
levels. Start at 0.6 V and stop at -0.1 V will achieve this. Once again
this does not change by going from COM to SEP i,e. by adding the 6 ft delay.

One other observation, probably related: After powerup it took several
vigorous switch activations of the COM/SEP switch before the thing swung
into action at all - i.e. the first reaction with COM setting was an Err
02, and with SEP I got the 15 ns as described above. So that switch may be
a bit dodgy.

Measuring in frequency mode gives the 10 MHz quite accurately, less than
1mHz off with its own ref as stop input. Feeding the 10 MHz in from my
GPSDO also ties up OK, it is around 40mHz off with perhaps 5-10mHz jitter.

Has anybody come across anything similar? I couldn't find much in that vein
in the archive but may have overlooked something. The one thing I was
worried about at first was the heat sinks at the rear getting really hot
but it seems I'm not the only one with that observation.

To some extent it looks like the input levels for triggering are off, but
then the fact that the 6 ft cable don't show up as extra delay may mean
there is something else wrong.

Looks like some "fun with Bill and Dave" lies ahead if somebody can point
me in the right direction.

Thanks in advance,
Thomas.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

I have a 5370B from ebay, it was missing a couple of boards and had a bad interpolator card. Also a bad input hybrid on the stop channel. My unit acts similar to yours, TI measurement on the 10mhz signal shows 107 ns. Frequency and Period measurement is spot on. I think it needs an alignment performed. I will get to that soon. Good Luck Gary - K5DSR -------- Forwarded Message -------- Subject: [time-nuts] 5370B Question / help needed Date: Sat, 1 Jul 2017 21:12:27 +0100 From: Thomas Allgeier <th.allgeier@gmail.com> Reply-To: Discussion of precise time and frequency measurement <time-nuts@febo.com> To: Discussion of precise time and frequency measurement <time-nuts@febo.com> I got hold of a 5370B cheaply but it turns out it may have a fault: Going through the motions of measuring TI with it's own 10MHz ref as input (as described in the manual) I don't get 100 ns, but around 15 ns. So this is with the switch set to START COM. Oddly enough I get the same 15 ns with the switch set to SEP, and going through a 6 ft cable between start and stop. Now I can "tune" the thing to display 100 ns by changing the trigger levels. Start at 0.6 V and stop at -0.1 V will achieve this. Once again this does not change by going from COM to SEP i,e. by adding the 6 ft delay. One other observation, probably related: After powerup it took several vigorous switch activations of the COM/SEP switch before the thing swung into action at all - i.e. the first reaction with COM setting was an Err 02, and with SEP I got the 15 ns as described above. So that switch may be a bit dodgy. Measuring in frequency mode gives the 10 MHz quite accurately, less than 1mHz off with its own ref as stop input. Feeding the 10 MHz in from my GPSDO also ties up OK, it is around 40mHz off with perhaps 5-10mHz jitter. Has anybody come across anything similar? I couldn't find much in that vein in the archive but may have overlooked something. The one thing I was worried about at first was the heat sinks at the rear getting really hot but it seems I'm not the only one with that observation. To some extent it looks like the input levels for triggering are off, but then the fact that the 6 ft cable don't show up as extra delay may mean there is something else wrong. Looks like some "fun with Bill and Dave" lies ahead if somebody can point me in the right direction. Thanks in advance, Thomas. _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
BK
Bob kb8tq
Mon, Jul 3, 2017 5:35 PM

Hi

If you go back in the archives a few years, there are extensive discussions of the various
alignment steps on a 5070A or B. They are a highly recommend read if you are getting
ready to dig into one of these counters.

Bob

On Jul 3, 2017, at 10:13 AM, Gary Neilson gary@deepskyridge.com wrote:

I have a 5370B from ebay, it was missing a couple of boards and had a bad interpolator card. Also a bad input hybrid on the stop channel.

My unit acts similar to yours, TI measurement on the 10mhz signal shows 107 ns.

Frequency and Period measurement is spot on.

I think it needs an alignment performed. I will get to that soon.

Good Luck

Gary - K5DSR

-------- Forwarded Message --------
Subject: [time-nuts] 5370B Question / help needed
Date: Sat, 1 Jul 2017 21:12:27 +0100
From: Thomas Allgeier th.allgeier@gmail.com
Reply-To: Discussion of precise time and frequency measurement time-nuts@febo.com
To: Discussion of precise time and frequency measurement time-nuts@febo.com

I got hold of a 5370B cheaply but it turns out it may have a fault:

Going through the motions of measuring TI with it's own 10MHz ref as input
(as described in the manual) I don't get 100 ns, but around 15 ns. So this
is with the switch set to START COM. Oddly enough I get the same 15 ns with
the switch set to SEP, and going through a 6 ft cable between start and
stop.

Now I can "tune" the thing to display 100 ns by changing the trigger
levels. Start at 0.6 V and stop at -0.1 V will achieve this. Once again
this does not change by going from COM to SEP i,e. by adding the 6 ft delay.

One other observation, probably related: After powerup it took several
vigorous switch activations of the COM/SEP switch before the thing swung
into action at all - i.e. the first reaction with COM setting was an Err
02, and with SEP I got the 15 ns as described above. So that switch may be
a bit dodgy.

Measuring in frequency mode gives the 10 MHz quite accurately, less than
1mHz off with its own ref as stop input. Feeding the 10 MHz in from my
GPSDO also ties up OK, it is around 40mHz off with perhaps 5-10mHz jitter.

Has anybody come across anything similar? I couldn't find much in that vein
in the archive but may have overlooked something. The one thing I was
worried about at first was the heat sinks at the rear getting really hot
but it seems I'm not the only one with that observation.

To some extent it looks like the input levels for triggering are off, but
then the fact that the 6 ft cable don't show up as extra delay may mean
there is something else wrong.

Looks like some "fun with Bill and Dave" lies ahead if somebody can point
me in the right direction.

Thanks in advance,
Thomas.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi If you go back in the archives a few years, there are extensive discussions of the various alignment steps on a 5070A or B. They are a highly recommend read if you are getting ready to dig into one of these counters. Bob > On Jul 3, 2017, at 10:13 AM, Gary Neilson <gary@deepskyridge.com> wrote: > > I have a 5370B from ebay, it was missing a couple of boards and had a bad interpolator card. Also a bad input hybrid on the stop channel. > > My unit acts similar to yours, TI measurement on the 10mhz signal shows 107 ns. > > Frequency and Period measurement is spot on. > > I think it needs an alignment performed. I will get to that soon. > > Good Luck > > Gary - K5DSR > > > > -------- Forwarded Message -------- > Subject: [time-nuts] 5370B Question / help needed > Date: Sat, 1 Jul 2017 21:12:27 +0100 > From: Thomas Allgeier <th.allgeier@gmail.com> > Reply-To: Discussion of precise time and frequency measurement <time-nuts@febo.com> > To: Discussion of precise time and frequency measurement <time-nuts@febo.com> > > > > I got hold of a 5370B cheaply but it turns out it may have a fault: > > Going through the motions of measuring TI with it's own 10MHz ref as input > (as described in the manual) I don't get 100 ns, but around 15 ns. So this > is with the switch set to START COM. Oddly enough I get the same 15 ns with > the switch set to SEP, and going through a 6 ft cable between start and > stop. > > Now I can "tune" the thing to display 100 ns by changing the trigger > levels. Start at 0.6 V and stop at -0.1 V will achieve this. Once again > this does not change by going from COM to SEP i,e. by adding the 6 ft delay. > > One other observation, probably related: After powerup it took several > vigorous switch activations of the COM/SEP switch before the thing swung > into action at all - i.e. the first reaction with COM setting was an Err > 02, and with SEP I got the 15 ns as described above. So that switch may be > a bit dodgy. > > Measuring in frequency mode gives the 10 MHz quite accurately, less than > 1mHz off with its own ref as stop input. Feeding the 10 MHz in from my > GPSDO also ties up OK, it is around 40mHz off with perhaps 5-10mHz jitter. > > Has anybody come across anything similar? I couldn't find much in that vein > in the archive but may have overlooked something. The one thing I was > worried about at first was the heat sinks at the rear getting really hot > but it seems I'm not the only one with that observation. > > To some extent it looks like the input levels for triggering are off, but > then the fact that the 6 ft cable don't show up as extra delay may mean > there is something else wrong. > > Looks like some "fun with Bill and Dave" lies ahead if somebody can point > me in the right direction. > > Thanks in advance, > Thomas. > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.