time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

Symmetricom/Datum FTS-1050A Disciplined Frequency Standard

EK
Erik Kaashoek
Wed, Jul 13, 2022 11:46 AM

Ed,
Thanks for the many good advice.
I've tried to incorporate as much as possible, updated schematic can be
found here: http://athome.kaashoek.com/time-nuts/PNA/SSPNA.JPG

For audio into the PC I'm using a professional balanced microphone to
USB input with a noise level of -130dBc/Hz and no spurs.
Using a 7805 and 1500uF capacitors I tried to create a solid reference
instead of the buffer op amp output and that did make a difference.
Further the input of the first opamp has been change to have identical
resistors at both the + and - input to reduce common mode signals.
None of the capacitors (5.6pF) I tried to reduce high frequency gain
improved the results. Most of the time it got worse.
Removal of a ceramic capacitors eliminated the microphony.

I've added a switch to select between 0dB and 20dB gain so I can
calibrate the level by offsetting the DUT frequency while keeping the
drive to the mixer.
To calibrate the effective noise BW of the FFT I create a test signal
combining a -70dBm 10.001MHz signal with a -90dBm/Hz noise signal.
The 20dB power ratio was confirmed using a calibrated spectrum analyzer.
The FFT length and sample rate at the PC where then changed till the PC
FFT showed the same power ratio.

Noise level at 1kHz is now -150dBc/Hz and -155dBc/Hz at 10kHz which is
the spec of the DOCXO used so no need to go any lower.
Also the 50Hz spurs and its harmonics are greatly reduced.

I'll have to invest in better coax cables as the current cable seem to
leak RF.
Erik.

Ed, Thanks for the many good advice. I've tried to incorporate as much as possible, updated schematic can be found here: http://athome.kaashoek.com/time-nuts/PNA/SSPNA.JPG For audio into the PC I'm using a professional balanced microphone to USB input with a noise level of -130dBc/Hz and no spurs. Using a 7805 and 1500uF capacitors I tried to create a solid reference instead of the buffer op amp output and that did make a difference. Further the input of the first opamp has been change to have identical resistors at both the + and - input to reduce common mode signals. None of the capacitors (5.6pF) I tried to reduce high frequency gain improved the results. Most of the time it got worse. Removal of a ceramic capacitors eliminated the microphony. I've added a switch to select between 0dB and 20dB gain so I can calibrate the level by offsetting the DUT frequency while keeping the drive to the mixer. To calibrate the effective noise BW of the FFT I create a test signal combining a -70dBm 10.001MHz signal with a -90dBm/Hz noise signal. The 20dB power ratio was confirmed using a calibrated spectrum analyzer. The FFT length and sample rate at the PC where then changed till the PC FFT showed the same power ratio. Noise level at 1kHz is now -150dBc/Hz and -155dBc/Hz at 10kHz which is the spec of the DOCXO used so no need to go any lower. Also the 50Hz spurs and its harmonics are greatly reduced. I'll have to invest in better coax cables as the current cable seem to leak RF. Erik.
BK
Bob kb8tq
Wed, Jul 13, 2022 4:54 PM

Hi

Well, obviously it’s now time to go get some better OCXO’s :) :) :)

One further note: It is tempting to go to multi section lowpass filtering
if you are trying to cover a range of frequencies. You also might try to push
the range of the noise measurement up to higher and higher offsets. There
is a point that all of this gets nutty. Peaking in the audio range is often the
net result. Spice is one way to watch out for it. Sweeping the circuit is
obviously a better way.

Bob

On Jul 13, 2022, at 3:46 AM, Erik Kaashoek via time-nuts time-nuts@lists.febo.com wrote:

Ed,
Thanks for the many good advice.
I've tried to incorporate as much as possible, updated schematic can be found here: http://athome.kaashoek.com/time-nuts/PNA/SSPNA.JPG

For audio into the PC I'm using a professional balanced microphone to USB input with a noise level of -130dBc/Hz and no spurs.
Using a 7805 and 1500uF capacitors I tried to create a solid reference instead of the buffer op amp output and that did make a difference.
Further the input of the first opamp has been change to have identical resistors at both the + and - input to reduce common mode signals.
None of the capacitors (5.6pF) I tried to reduce high frequency gain improved the results. Most of the time it got worse.
Removal of a ceramic capacitors eliminated the microphony.

I've added a switch to select between 0dB and 20dB gain so I can calibrate the level by offsetting the DUT frequency while keeping the drive to the mixer.
To calibrate the effective noise BW of the FFT I create a test signal combining a -70dBm 10.001MHz signal with a -90dBm/Hz noise signal.
The 20dB power ratio was confirmed using a calibrated spectrum analyzer. The FFT length and sample rate at the PC where then changed till the PC FFT showed the same power ratio.

Noise level at 1kHz is now -150dBc/Hz and -155dBc/Hz at 10kHz which is the spec of the DOCXO used so no need to go any lower.
Also the 50Hz spurs and its harmonics are greatly reduced.

I'll have to invest in better coax cables as the current cable seem to leak RF.
Erik.


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

Hi Well, obviously it’s now time to go get some better OCXO’s :) :) :) One further note: It is tempting to go to multi section lowpass filtering if you are trying to cover a range of frequencies. You also might try to push the range of the noise measurement up to higher and higher offsets. There is a point that all of this gets nutty. Peaking in the audio range is often the net result. Spice is one way to watch out for it. Sweeping the circuit is obviously a better way. Bob > On Jul 13, 2022, at 3:46 AM, Erik Kaashoek via time-nuts <time-nuts@lists.febo.com> wrote: > > Ed, > Thanks for the many good advice. > I've tried to incorporate as much as possible, updated schematic can be found here: http://athome.kaashoek.com/time-nuts/PNA/SSPNA.JPG > > For audio into the PC I'm using a professional balanced microphone to USB input with a noise level of -130dBc/Hz and no spurs. > Using a 7805 and 1500uF capacitors I tried to create a solid reference instead of the buffer op amp output and that did make a difference. > Further the input of the first opamp has been change to have identical resistors at both the + and - input to reduce common mode signals. > None of the capacitors (5.6pF) I tried to reduce high frequency gain improved the results. Most of the time it got worse. > Removal of a ceramic capacitors eliminated the microphony. > > I've added a switch to select between 0dB and 20dB gain so I can calibrate the level by offsetting the DUT frequency while keeping the drive to the mixer. > To calibrate the effective noise BW of the FFT I create a test signal combining a -70dBm 10.001MHz signal with a -90dBm/Hz noise signal. > The 20dB power ratio was confirmed using a calibrated spectrum analyzer. The FFT length and sample rate at the PC where then changed till the PC FFT showed the same power ratio. > > Noise level at 1kHz is now -150dBc/Hz and -155dBc/Hz at 10kHz which is the spec of the DOCXO used so no need to go any lower. > Also the 50Hz spurs and its harmonics are greatly reduced. > > I'll have to invest in better coax cables as the current cable seem to leak RF. > Erik. > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com
EK
Erik Kaashoek
Wed, Jul 13, 2022 5:30 PM

The oscillator I wanted to measure with the PNA (a cheap Chines TCXO)
can now be measured with 30dB margin and the internal DOCXO is better
than any other oscillator I have so I could say I'm done, apart from
becoming a phase noise nut :-)
The whole build is still Manhattan style with long component leads so
making a more compact build on a solid copper PCB may further improve
performance.
@Ed:  The 7805 was temporarily replaced with a linear bench supply which
made no visible difference
@Bob: The audio input I'm using claims to be 192kHz but in reality it
stops at 96kHz so the upper limit is 48kHz for now, which is good
enough. The calibration method you advised  (offset DUT) works very well
now the gain is switchable.
Erik.

On 13-7-2022 18:54, Bob kb8tq wrote:

Hi

Well, obviously it’s now time to go get some better OCXO’s :) :) :)

One further note: It is tempting to go to multi section lowpass filtering
if you are trying to cover a range of frequencies. You also might try to push
the range of the noise measurement up to higher and higher offsets. There
is a point that all of this gets nutty. Peaking in the audio range is often the
net result. Spice is one way to watch out for it. Sweeping the circuit is
obviously a better way.

Bob

The oscillator I wanted to measure with the PNA (a cheap Chines TCXO) can now be measured with 30dB margin and the internal DOCXO is better than any other oscillator I have so I could say I'm done, apart from becoming a phase noise nut :-) The whole build is still Manhattan style with long component leads so making a more compact build on a solid copper PCB may further improve performance. @Ed:  The 7805 was temporarily replaced with a linear bench supply which made no visible difference @Bob: The audio input I'm using claims to be 192kHz but in reality it stops at 96kHz so the upper limit is 48kHz for now, which is good enough. The calibration method you advised  (offset DUT) works very well now the gain is switchable. Erik. On 13-7-2022 18:54, Bob kb8tq wrote: > Hi > > Well, obviously it’s now time to go get some better OCXO’s :) :) :) > > One further note: It is tempting to go to multi section lowpass filtering > if you are trying to cover a range of frequencies. You also might try to push > the range of the noise measurement up to higher and higher offsets. There > is a point that all of this gets nutty. Peaking in the audio range is often the > net result. Spice is one way to watch out for it. Sweeping the circuit is > obviously a better way. > > Bob >
BK
Bob kb8tq
Wed, Jul 13, 2022 6:36 PM

Hi

One of the “interesting” gotchas with some sound cards:

The frequency response goes to (say) 48 KHz. The noise
performance goes to (maybe) 20 KHz. Past that point there
is a nasty rise in the noise floor. How common this is … who
knows.

For a lot of applications, the system floors out well before
10 KHz. There really isn’t a lot of need for phase noise past
that ( other than for spurs ).

It’s debatable just what the “best” approach for wide band
spurs is. A notch filter in front of a spectrum analyzer is a
often a good way to go.

Bob

On Jul 13, 2022, at 9:30 AM, Erik Kaashoek erik@kaashoek.com wrote:

The oscillator I wanted to measure with the PNA (a cheap Chines TCXO) can now be measured with 30dB margin and the internal DOCXO is better than any other oscillator I have so I could say I'm done, apart from becoming a phase noise nut :-)
The whole build is still Manhattan style with long component leads so making a more compact build on a solid copper PCB may further improve performance.
@Ed:  The 7805 was temporarily replaced with a linear bench supply which made no visible difference
@Bob: The audio input I'm using claims to be 192kHz but in reality it stops at 96kHz so the upper limit is 48kHz for now, which is good enough. The calibration method you advised  (offset DUT) works very well now the gain is switchable.
Erik.

On 13-7-2022 18:54, Bob kb8tq wrote:

Hi

Well, obviously it’s now time to go get some better OCXO’s :) :) :)

One further note: It is tempting to go to multi section lowpass filtering
if you are trying to cover a range of frequencies. You also might try to push
the range of the noise measurement up to higher and higher offsets. There
is a point that all of this gets nutty. Peaking in the audio range is often the
net result. Spice is one way to watch out for it. Sweeping the circuit is
obviously a better way.

Bob

Hi One of the “interesting” gotchas with some sound cards: The frequency response goes to (say) 48 KHz. The noise performance goes to (maybe) 20 KHz. Past that point there is a nasty rise in the noise floor. How common this is … who knows. For a lot of applications, the system floors out well before 10 KHz. There really isn’t a lot of need for phase noise past that ( other than for spurs ). It’s debatable just what the “best” approach for wide band spurs is. A notch filter in front of a spectrum analyzer is a often a good way to go. Bob > On Jul 13, 2022, at 9:30 AM, Erik Kaashoek <erik@kaashoek.com> wrote: > > The oscillator I wanted to measure with the PNA (a cheap Chines TCXO) can now be measured with 30dB margin and the internal DOCXO is better than any other oscillator I have so I could say I'm done, apart from becoming a phase noise nut :-) > The whole build is still Manhattan style with long component leads so making a more compact build on a solid copper PCB may further improve performance. > @Ed: The 7805 was temporarily replaced with a linear bench supply which made no visible difference > @Bob: The audio input I'm using claims to be 192kHz but in reality it stops at 96kHz so the upper limit is 48kHz for now, which is good enough. The calibration method you advised (offset DUT) works very well now the gain is switchable. > Erik. > > On 13-7-2022 18:54, Bob kb8tq wrote: >> Hi >> >> Well, obviously it’s now time to go get some better OCXO’s :) :) :) >> >> One further note: It is tempting to go to multi section lowpass filtering >> if you are trying to cover a range of frequencies. You also might try to push >> the range of the noise measurement up to higher and higher offsets. There >> is a point that all of this gets nutty. Peaking in the audio range is often the >> net result. Spice is one way to watch out for it. Sweeping the circuit is >> obviously a better way. >> >> Bob >> >
EB
ed breya
Wed, Jul 13, 2022 8:38 PM

Erik, that looks pretty nice, with good stout bypassing. I'd recommend
adding some ceramic caps bypassing for RF, just in case. Also, I believe
the standard three-terminal regulators have no limit on capacitive
loading, so you probably don't need R1. The standard 7805 can put out a
lot of current (>1A), so I'd suggest using the 78L05 low power version
(100 mA or so). That will make it a little safer, fault-wise. Especially
when building and experimenting, it's good to have some limiting
appropriate to the circuit loads. If you should say, inadvertently short
the circuit common to ground, a way bigger than needed regulator will
supply lots of current, and possibly crash the +12V if it can't keep up.
Finally, whenever these regulators have large output capacitors, it's
good to have a reverse protection diode from output to input, in case
the input voltage is suddenly dumped - another thing that can easily happen.

So, with good DC and LF regulation, and plenty of audio and RF
bypassing, the common line should be very solid. I don't think the very
low frequency regulator noise or DC drift will cause any grief, since
the opamp's CMRR should take care of it. I looked up the OP-27, and it
shows over 120 dB CMRR up to a few kHz, so the CM to input signal
conversion is less than 1 uV/V. You should check all the part specs to
be sure if it will be OK. The PSRR is much weaker, so the cleanliness of
the +12V supply should be considered too.

I see that you've changed the LNA circuit (to inverting) and feed to the
audio (ground-referenced - nice). Also, it appears that 10X voltage gain
is sufficient, instead of the 100X previous. This should greatly reduce
the grief. One thing that's different about the inverting mode versus
the previous non-inverting, is that now you have a definite DC/LF
termination/load resistance (the 100 ohms into the summing node) on the
mixer IF port, while the non-inverting mode can provide very high input
R, so you can have more flexibility on termination R choice, from "open"
on down to whatever you want. I don't know what the ideal is, but as in
previous discussions, it doesn't need to be 50 ohms as in RF work, but
can be much higher for the audio, to get a bigger voltage signal. For
inverting mode, the input R must be all or part of the termination R,
and "open" is not an option.

I'm wondering what the PC USB audio box is. Could you please let us know
make, model, etc? The "balanced microphone" description sounds like it's
600 ohm input R.

Ed

Erik, that looks pretty nice, with good stout bypassing. I'd recommend adding some ceramic caps bypassing for RF, just in case. Also, I believe the standard three-terminal regulators have no limit on capacitive loading, so you probably don't need R1. The standard 7805 can put out a lot of current (>1A), so I'd suggest using the 78L05 low power version (100 mA or so). That will make it a little safer, fault-wise. Especially when building and experimenting, it's good to have some limiting appropriate to the circuit loads. If you should say, inadvertently short the circuit common to ground, a way bigger than needed regulator will supply lots of current, and possibly crash the +12V if it can't keep up. Finally, whenever these regulators have large output capacitors, it's good to have a reverse protection diode from output to input, in case the input voltage is suddenly dumped - another thing that can easily happen. So, with good DC and LF regulation, and plenty of audio and RF bypassing, the common line should be very solid. I don't think the very low frequency regulator noise or DC drift will cause any grief, since the opamp's CMRR should take care of it. I looked up the OP-27, and it shows over 120 dB CMRR up to a few kHz, so the CM to input signal conversion is less than 1 uV/V. You should check all the part specs to be sure if it will be OK. The PSRR is much weaker, so the cleanliness of the +12V supply should be considered too. I see that you've changed the LNA circuit (to inverting) and feed to the audio (ground-referenced - nice). Also, it appears that 10X voltage gain is sufficient, instead of the 100X previous. This should greatly reduce the grief. One thing that's different about the inverting mode versus the previous non-inverting, is that now you have a definite DC/LF termination/load resistance (the 100 ohms into the summing node) on the mixer IF port, while the non-inverting mode can provide very high input R, so you can have more flexibility on termination R choice, from "open" on down to whatever you want. I don't know what the ideal is, but as in previous discussions, it doesn't need to be 50 ohms as in RF work, but can be much higher for the audio, to get a bigger voltage signal. For inverting mode, the input R must be all or part of the termination R, and "open" is not an option. I'm wondering what the PC USB audio box is. Could you please let us know make, model, etc? The "balanced microphone" description sounds like it's 600 ohm input R. Ed
GE
glen english LIST
Wed, Jul 13, 2022 11:10 PM

Most of those are now delta signma converters, the noise is moved from
the baseband  to 'elsewhere'.

be careful with levels of  out of band (beyond designed freq response
say 20kHz) because many of the AFEs are 'economically designed ' and
slew rate overload of 'loud' super nyquist energy seems to be a bit of
an issue.

Apart from 1/f  measurement, two tone IMD , a useful measurement to
ascertain the performance of the converter in a high broadband noise
environment is a Noise Power Ratio Test. Basically , proceeded by a
broadband noise source, a notch is used, and what is left over in the
notch (should be nothing) . It's a great test.

This article talks about ADCs at lower HF, but the technique is equally
at home for audio band ADCs

https://www.ab4oj.com/test/docs/npr_test.pdf

Adam has written a good article.

-Glen.

On 14/07/2022 4:36 am, Bob kb8tq via time-nuts wrote:

Hi

One of the “interesting” gotchas with some sound cards:

The frequency response goes to (say) 48 KHz. The noise
performance goes to (maybe) 20 KHz. Past that point there
is a nasty rise in the noise floor. How common this is … who
knows.

For a lot of applications, the system floors out well before
10 KHz. There really isn’t a lot of need for phase noise past
that ( other than for spurs ).

It’s debatable just what the “best” approach for wide band

Most of those are now delta signma converters, the noise is moved from the baseband  to 'elsewhere'. be careful with levels of  out of band (beyond designed freq response say 20kHz) because many of the AFEs are 'economically designed ' and slew rate overload of 'loud' super nyquist energy seems to be a bit of an issue. Apart from 1/f  measurement, two tone IMD , a useful measurement to ascertain the performance of the converter in a high broadband noise environment is a Noise Power Ratio Test. Basically , proceeded by a broadband noise source, a notch is used, and what is left over in the notch (should be nothing) . It's a great test. This article talks about ADCs at lower HF, but the technique is equally at home for audio band ADCs https://www.ab4oj.com/test/docs/npr_test.pdf Adam has written a good article. -Glen. On 14/07/2022 4:36 am, Bob kb8tq via time-nuts wrote: > Hi > > One of the “interesting” gotchas with some sound cards: > > The frequency response goes to (say) 48 KHz. The noise > performance goes to (maybe) 20 KHz. Past that point there > is a nasty rise in the noise floor. How common this is … who > knows. > > For a lot of applications, the system floors out well before > 10 KHz. There really isn’t a lot of need for phase noise past > that ( other than for spurs ). > > It’s debatable just what the “best” approach for wide band
EK
Erik Kaashoek
Thu, Jul 14, 2022 10:54 AM

Ed,
I'm using a Behringer UMC202HD
https://www.behringer.com/product.html?modelCode=P0BJZ
The noise floor between 100Hz and 50kHz is about -140dBc/Hz and it is
almost flat down to 1Hz.
Erik.

On 13-7-2022 22:38, ed breya via time-nuts wrote:

I'm wondering what the PC USB audio box is. Could you please let us
know make, model, etc? The "balanced microphone" description sounds
like it's 600 ohm input R.

Ed, I'm using a Behringer UMC202HD https://www.behringer.com/product.html?modelCode=P0BJZ The noise floor between 100Hz and 50kHz is about -140dBc/Hz and it is almost flat down to 1Hz. Erik. On 13-7-2022 22:38, ed breya via time-nuts wrote: > I'm wondering what the PC USB audio box is. Could you please let us > know make, model, etc? The "balanced microphone" description sounds > like it's 600 ohm input R.
EB
ed breya
Fri, Jul 15, 2022 8:37 PM

Erik, it sounds like you have it nearly finished and working - congrats.
You should be able to greatly reduce those pesky line-interference
spurs. The first weak link may be the audio signal path, so it's time to
let the audio box help, if you haven't already optimized it. In the last
design round, you now have the audio box driven from a coax line with
its shield connected to your local instrumentation ground, which is
earth ground there, and the minus end of the 12V supply, right?

How is the other end hooked up to the audio box? I looked at some info
on the unit, but it doesn't say anything about hookup details - a
"normal" user (not us) would probably just plug in a mic and go. If it
is a balanced mic connection, then it is at least some form of
differential input, and you had been using it this way before, with the
audio common (low side) connected to your floating signal common. That's
the way you should connect the audio box still, even though the cable is
now earth ground referenced at the the instrument box. If it's still set
up this way anyway, then OK. If you have connected the cable shield to
the ground at the PC or USB audio box, try removing it. The audio box
input should have some amount of LF CMR, which should help in this part.
You can experiment with different arrangements and see what works best.
The next thing is to try adding common-mode choking with various methods

  • I'll explain later.

From the very beginning, I think, you have considered the benefits of
galvanic isolation of the DBM/PD mixer ports, which is good.
Unfortunately, powering up the OCXOs and getting their signals connected
leads to some inevitable ground loops anyway, so you can't always take
full advantage of built-in isolation, but at least you have it. Try to
use it when you can.

The rest of the line and low frequency ground loop issues are hard to
predict, depending on the overall powering and operational setups, but
some general methods can be applied.

Now back to the audio output path, the first thing is to see if there's
any point to improving its CMR - if the interference is getting in
elsewhere, it doesn't matter (yet) how good it is. You can experiment
with the circuit hookup and operation. For instance, in your latest
post, the first chart shows the operating noise floor and line spurs
present "without the DUT connected." This can describe a range of
possibilities, from the RF connector being unplugged, to the DUT being
totally removed from the experiment, power-wise, ground-wise, and so on.
In your setup, you have a number of items hooked up, each contributing
to the ground loop situation.

You can apply the process of elimination to gradually improve it. The
first and easiest step is to shut everything down except the PC and
audio analyzer setup, and assess the spurs. If there's little change,
then the interference is what it is, due to the environment of the
setup. If the spurs go away, then your analyzer circuit is amplifying
them from the signal chain, or from the power supplies. Then you would
short the input of the LNA, power it up and look at the spurs again, and
so on.

This narrows things down as you go, with various tests to see what
things make the biggest difference. If the spurs are relatively
unchanged with power down, then you would unhook things until you're
left with just the analyzer circuit hanging from the cable to the audio
box - you systematically delete connections to power supplies, grounds,
signals etc, until it shows clean. Then you can add back and think about
the possible ways the interference gets in.

Once you figure out the major interference sources, you'll know what and
where to make improvements. Keep in mind that within your own project
circuit, you have lots of options, since you're building it, while on
most of the other stuff, you'll likely be limited to external fixes like
CM chokes (prayer beads) and grounding tricks.

Next time I'll talk about some of the methods in detail.

Ed

Erik, it sounds like you have it nearly finished and working - congrats. You should be able to greatly reduce those pesky line-interference spurs. The first weak link may be the audio signal path, so it's time to let the audio box help, if you haven't already optimized it. In the last design round, you now have the audio box driven from a coax line with its shield connected to your local instrumentation ground, which is earth ground there, and the minus end of the 12V supply, right? How is the other end hooked up to the audio box? I looked at some info on the unit, but it doesn't say anything about hookup details - a "normal" user (not us) would probably just plug in a mic and go. If it is a balanced mic connection, then it is at least some form of differential input, and you had been using it this way before, with the audio common (low side) connected to your floating signal common. That's the way you should connect the audio box still, even though the cable is now earth ground referenced at the the instrument box. If it's still set up this way anyway, then OK. If you have connected the cable shield to the ground at the PC or USB audio box, try removing it. The audio box input should have some amount of LF CMR, which should help in this part. You can experiment with different arrangements and see what works best. The next thing is to try adding common-mode choking with various methods - I'll explain later. From the very beginning, I think, you have considered the benefits of galvanic isolation of the DBM/PD mixer ports, which is good. Unfortunately, powering up the OCXOs and getting their signals connected leads to some inevitable ground loops anyway, so you can't always take full advantage of built-in isolation, but at least you have it. Try to use it when you can. The rest of the line and low frequency ground loop issues are hard to predict, depending on the overall powering and operational setups, but some general methods can be applied. Now back to the audio output path, the first thing is to see if there's any point to improving its CMR - if the interference is getting in elsewhere, it doesn't matter (yet) how good it is. You can experiment with the circuit hookup and operation. For instance, in your latest post, the first chart shows the operating noise floor and line spurs present "without the DUT connected." This can describe a range of possibilities, from the RF connector being unplugged, to the DUT being totally removed from the experiment, power-wise, ground-wise, and so on. In your setup, you have a number of items hooked up, each contributing to the ground loop situation. You can apply the process of elimination to gradually improve it. The first and easiest step is to shut everything down except the PC and audio analyzer setup, and assess the spurs. If there's little change, then the interference is what it is, due to the environment of the setup. If the spurs go away, then your analyzer circuit is amplifying them from the signal chain, or from the power supplies. Then you would short the input of the LNA, power it up and look at the spurs again, and so on. This narrows things down as you go, with various tests to see what things make the biggest difference. If the spurs are relatively unchanged with power down, then you would unhook things until you're left with just the analyzer circuit hanging from the cable to the audio box - you systematically delete connections to power supplies, grounds, signals etc, until it shows clean. Then you can add back and think about the possible ways the interference gets in. Once you figure out the major interference sources, you'll know what and where to make improvements. Keep in mind that within your own project circuit, you have lots of options, since you're building it, while on most of the other stuff, you'll likely be limited to external fixes like CM chokes (prayer beads) and grounding tricks. Next time I'll talk about some of the methods in detail. Ed
DD
Dr. David Kirkby
Fri, Jul 15, 2022 11:53 PM

On Fri, 15 Jul 2022 at 23:35, ed breya via time-nuts <
time-nuts@lists.febo.com> wrote:

Unfortunately, powering up the OCXOs and getting their signals connected
leads to some inevitable ground loops anyway, so you can't always take
full advantage of built-in isolation, but at least you have it. Try to
use it when you can.

The rest of the line and low frequency ground loop issues are hard to
predict, depending on the overall powering and operational setups, but
some general methods can be applied.

Would it not be worth running from NiCd batteries to eliminate PSU noise
and conducted interference as a possible source. I believe NiCd have one of
the lowest noises, bit no doubt Google would help.

I have a couple of high resistance meters (ie basically electrometers). A
lot of the signals in that are optically coupled between boards.

Dave

Dr. David Kirkby,
Kirkby Microwave Ltd,
drkirkby@kirkbymicrowave.co.uk
https://www.kirkbymicrowave.co.uk/
Telephone 01621-680100./ +44 1621 680100

Registered in England & Wales, company number 08914892.
Registered office:
Stokes Hall Lodge, Burnham Rd, Althorne, Chelmsford, Essex, CM3 6DT, United
Kingdom

On Fri, 15 Jul 2022 at 23:35, ed breya via time-nuts < time-nuts@lists.febo.com> wrote: > > Unfortunately, powering up the OCXOs and getting their signals connected > leads to some inevitable ground loops anyway, so you can't always take > full advantage of built-in isolation, but at least you have it. Try to > use it when you can. > > The rest of the line and low frequency ground loop issues are hard to > predict, depending on the overall powering and operational setups, but > some general methods can be applied. > Would it not be worth running from NiCd batteries to eliminate PSU noise and conducted interference as a possible source. I believe NiCd have one of the lowest noises, bit no doubt Google would help. I have a couple of high resistance meters (ie basically electrometers). A lot of the signals in that are optically coupled between boards. Dave -- Dr. David Kirkby, Kirkby Microwave Ltd, drkirkby@kirkbymicrowave.co.uk https://www.kirkbymicrowave.co.uk/ Telephone 01621-680100./ +44 1621 680100 Registered in England & Wales, company number 08914892. Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, Chelmsford, Essex, CM3 6DT, United Kingdom
BK
Bob kb8tq
Sat, Jul 16, 2022 1:44 AM

Hi

Setting a battery output to some exact voltage ( say 12.000 V ) is a bit exciting.
OCXO performance in many cases is voltage dependent. You do want to run
them at a constant / correct voltage as you characterize them.

Bob

On Jul 15, 2022, at 3:53 PM, Dr. David Kirkby via time-nuts time-nuts@lists.febo.com wrote:

On Fri, 15 Jul 2022 at 23:35, ed breya via time-nuts <
time-nuts@lists.febo.com> wrote:

Unfortunately, powering up the OCXOs and getting their signals connected
leads to some inevitable ground loops anyway, so you can't always take
full advantage of built-in isolation, but at least you have it. Try to
use it when you can.

The rest of the line and low frequency ground loop issues are hard to
predict, depending on the overall powering and operational setups, but
some general methods can be applied.

Would it not be worth running from NiCd batteries to eliminate PSU noise
and conducted interference as a possible source. I believe NiCd have one of
the lowest noises, bit no doubt Google would help.

I have a couple of high resistance meters (ie basically electrometers). A
lot of the signals in that are optically coupled between boards.

Dave

Dr. David Kirkby,
Kirkby Microwave Ltd,
drkirkby@kirkbymicrowave.co.uk
https://www.kirkbymicrowave.co.uk/
Telephone 01621-680100./ +44 1621 680100

Registered in England & Wales, company number 08914892.
Registered office:
Stokes Hall Lodge, Burnham Rd, Althorne, Chelmsford, Essex, CM3 6DT, United
Kingdom


time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe send an email to time-nuts-leave@lists.febo.com

Hi Setting a battery output to some exact voltage ( say 12.000 V ) is a bit exciting. OCXO performance in many cases *is* voltage dependent. You do want to run them at a constant / correct voltage as you characterize them. Bob > On Jul 15, 2022, at 3:53 PM, Dr. David Kirkby via time-nuts <time-nuts@lists.febo.com> wrote: > > On Fri, 15 Jul 2022 at 23:35, ed breya via time-nuts < > time-nuts@lists.febo.com> wrote: > >> >> Unfortunately, powering up the OCXOs and getting their signals connected >> leads to some inevitable ground loops anyway, so you can't always take >> full advantage of built-in isolation, but at least you have it. Try to >> use it when you can. >> >> The rest of the line and low frequency ground loop issues are hard to >> predict, depending on the overall powering and operational setups, but >> some general methods can be applied. >> > > Would it not be worth running from NiCd batteries to eliminate PSU noise > and conducted interference as a possible source. I believe NiCd have one of > the lowest noises, bit no doubt Google would help. > > I have a couple of high resistance meters (ie basically electrometers). A > lot of the signals in that are optically coupled between boards. > > Dave > -- > Dr. David Kirkby, > Kirkby Microwave Ltd, > drkirkby@kirkbymicrowave.co.uk > https://www.kirkbymicrowave.co.uk/ > Telephone 01621-680100./ +44 1621 680100 > > Registered in England & Wales, company number 08914892. > Registered office: > Stokes Hall Lodge, Burnham Rd, Althorne, Chelmsford, Essex, CM3 6DT, United > Kingdom > _______________________________________________ > time-nuts mailing list -- time-nuts@lists.febo.com > To unsubscribe send an email to time-nuts-leave@lists.febo.com