Hi,

My current Tally of  "Known" HP 5065A is 120 units!

Cheers,

Corby

Hi, My current Tally of "Known" HP 5065A is 120 units! Cheers, Corby
PL
Pete Lancashire
Sat, Feb 24, 2018 3:25 AM

Wow, don't know why but more than I expected. I no longer do, but if the
person who I gave it to wants and is on TN can.

-pet

On Fri, Feb 23, 2018 at 6:23 PM, cdelect@juno.com wrote:

Hi,

My current Tally of  "Known" HP 5065A is 120 units!

Cheers,

Corby


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

Wow, don't know why but more than I expected. I no longer do, but if the person who I gave it to wants and is on TN can. -pet On Fri, Feb 23, 2018 at 6:23 PM, <cdelect@juno.com> wrote: > Hi, > > My current Tally of "Known" HP 5065A is 120 units! > > Cheers, > > Corby > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > and follow the instructions there. >
BK
Bob kb8tq
Sat, Feb 24, 2018 3:34 AM

Hi

Well, if anybody else is “giving away” 5065’s I’d certainly be willing to “accept”
one :)

Bob

On Feb 23, 2018, at 10:25 PM, Pete Lancashire pete@petelancashire.com wrote:

Wow, don't know why but more than I expected. I no longer do, but if the
person who I gave it to wants and is on TN can.

-pet

On Fri, Feb 23, 2018 at 6:23 PM, cdelect@juno.com wrote:

Hi,

My current Tally of  "Known" HP 5065A is 120 units!

Cheers,

Corby


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Well, if anybody else is “giving away” 5065’s I’d certainly be willing to “accept” one :) Bob > On Feb 23, 2018, at 10:25 PM, Pete Lancashire <pete@petelancashire.com> wrote: > > Wow, don't know why but more than I expected. I no longer do, but if the > person who I gave it to wants and is on TN can. > > -pet > > On Fri, Feb 23, 2018 at 6:23 PM, <cdelect@juno.com> wrote: > >> Hi, >> >> My current Tally of "Known" HP 5065A is 120 units! >> >> Cheers, >> >> Corby >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/ >> mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
JM
John Miles
Sat, Feb 24, 2018 10:02 AM

Hi

Well, if anybody else is “giving away” 5065’s I’d certainly be willing to
“accept”
one :)

Bob

Be careful what you wish for. :) The unit Pete is talking about is a 1969-era model, in nice overall shape but with the usual bad capacitors on the oven controller board.  The lamp oven winding should be about 50 ohms cold, but is closer to a dead short.  It will take a LOT of work to restore to working condition.

The failed controller ran long enough to smoke a 1.5-ohm 1W resistor, 1N400-something diode, and eventually the 1A line fuse ( http://www.ke5fx.com/5065A_A11_sm.jpg ).  Needless to say the lamp PCB looks like something out of Fukushima.  It'll need to be rebuilt from scratch after rewinding the heater.

Seriously -- anyone with a 5065A who hasn't checked/replaced the caps on A11, don't let this happen to you.  They aren't making any more of these puppies.  If the caps on that board are original, don't bother to check them, just replace them, as Luciano suggests at http://www.timeok.it/hp5065a-corner-3/.  No need for exotic parts, just put in whatever you have that is somewhere close to the original values.

Not a bad idea to verify the ESR of the new parts you're installing as well.

-- john, KE5FX
Miles Design LLC

> Hi > > Well, if anybody else is “giving away” 5065’s I’d certainly be willing to > “accept” > one :) > > Bob Be careful what you wish for. :) The unit Pete is talking about is a 1969-era model, in nice overall shape but with the usual bad capacitors on the oven controller board. The lamp oven winding should be about 50 ohms cold, but is closer to a dead short. It will take a LOT of work to restore to working condition. The failed controller ran long enough to smoke a 1.5-ohm 1W resistor, 1N400-something diode, and eventually the 1A line fuse ( http://www.ke5fx.com/5065A_A11_sm.jpg ). Needless to say the lamp PCB looks like something out of Fukushima. It'll need to be rebuilt from scratch after rewinding the heater. Seriously -- anyone with a 5065A who hasn't checked/replaced the caps on A11, don't let this happen to you. They aren't making any more of these puppies. If the caps on that board are original, don't bother to check them, just replace them, as Luciano suggests at http://www.timeok.it/hp5065a-corner-3/. No need for exotic parts, just put in whatever you have that is somewhere close to the original values. Not a bad idea to verify the ESR of the new parts you're installing as well. -- john, KE5FX Miles Design LLC
GA
George Atkinson
Sat, Feb 24, 2018 10:46 AM

There is one on UK ebay at the moment but its not being given away. From the one partial photograph it looks a bit rough.

https://www.ebay.co.uk/itm/HP-5065A-Rubidium-Frequency-Standard/183057846558?

Robert G8RPI.

 On 24 February 2018 at 10:02 John Miles <john@miles.io> wrote:
     Hi

     Well, if anybody else is “giving away” 5065’s I’d certainly be willing to
     “accept”
     one :)

     Bob
 Be careful what you wish for. :) The unit Pete is talking about is a 1969-era model, in nice overall shape but with the usual bad capacitors on the oven controller board. The lamp oven winding should be about 50 ohms cold, but is closer to a dead short. It will take a LOT of work to restore to working condition.

 The failed controller ran long enough to smoke a 1.5-ohm 1W resistor, 1N400-something diode, and eventually the 1A line fuse ( http://www.ke5fx.com/5065A_A11_sm.jpg ). Needless to say the lamp PCB looks like something out of Fukushima. It'll need to be rebuilt from scratch after rewinding the heater.

 Seriously -- anyone with a 5065A who hasn't checked/replaced the caps on A11, don't let this happen to you. They aren't making any more of these puppies. If the caps on that board are original, don't bother to check them, just replace them, as Luciano suggests at http://www.timeok.it/hp5065a-corner-3/. No need for exotic parts, just put in whatever you have that is somewhere close to the original values.

 Not a bad idea to verify the ESR of the new parts you're installing as well.

 -- john, KE5FX
 Miles Design LLC

 _______________________________________________
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
There is one on UK ebay at the moment but its not being given away. From the one partial photograph it looks a bit rough. https://www.ebay.co.uk/itm/HP-5065A-Rubidium-Frequency-Standard/183057846558? Robert G8RPI. > > On 24 February 2018 at 10:02 John Miles <john@miles.io> wrote: > > > > > > Hi > > > > Well, if anybody else is “giving away” 5065’s I’d certainly be willing to > > “accept” > > one :) > > > > Bob > > > > > > Be careful what you wish for. :) The unit Pete is talking about is a 1969-era model, in nice overall shape but with the usual bad capacitors on the oven controller board. The lamp oven winding should be about 50 ohms cold, but is closer to a dead short. It will take a LOT of work to restore to working condition. > > The failed controller ran long enough to smoke a 1.5-ohm 1W resistor, 1N400-something diode, and eventually the 1A line fuse ( http://www.ke5fx.com/5065A_A11_sm.jpg ). Needless to say the lamp PCB looks like something out of Fukushima. It'll need to be rebuilt from scratch after rewinding the heater. > > Seriously -- anyone with a 5065A who hasn't checked/replaced the caps on A11, don't let this happen to you. They aren't making any more of these puppies. If the caps on that board are original, don't bother to check them, just replace them, as Luciano suggests at http://www.timeok.it/hp5065a-corner-3/. No need for exotic parts, just put in whatever you have that is somewhere close to the original values. > > Not a bad idea to verify the ESR of the new parts you're installing as well. > > -- john, KE5FX > Miles Design LLC > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
CJ
Clint Jay
Sat, Feb 24, 2018 11:12 AM

It looks rough, there's a cesium standard for sale in the UK, time nuts
member perhaps, that's sat at a quarter of the price, I know which I'd buy.

On 24 Feb 2018 10:46, "George Atkinson via time-nuts" time-nuts@febo.com
wrote:

There is one on UK ebay at the moment but its not being given away. From
the one partial photograph it looks a bit rough.

https://www.ebay.co.uk/itm/HP-5065A-Rubidium-Frequency-
Standard/183057846558?

Robert G8RPI.

 On 24 February 2018 at 10:02 John Miles <john@miles.io> wrote:
     Hi

     Well, if anybody else is “giving away” 5065’s I’d certainly be

willing to

     “accept”
     one :)

     Bob
 Be careful what you wish for. :) The unit Pete is talking about is a

1969-era model, in nice overall shape but with the usual bad capacitors on
the oven controller board. The lamp oven winding should be about 50 ohms
cold, but is closer to a dead short. It will take a LOT of work to restore
to working condition.

 The failed controller ran long enough to smoke a 1.5-ohm 1W

resistor, 1N400-something diode, and eventually the 1A line fuse (
http://www.ke5fx.com/5065A_A11_sm.jpg ). Needless to say the lamp PCB
looks like something out of Fukushima. It'll need to be rebuilt from
scratch after rewinding the heater.

 Seriously -- anyone with a 5065A who hasn't checked/replaced the

caps on A11, don't let this happen to you. They aren't making any more of
these puppies. If the caps on that board are original, don't bother to
check them, just replace them, as Luciano suggests at
http://www.timeok.it/hp5065a-corner-3/. No need for exotic parts, just
put in whatever you have that is somewhere close to the original values.

 Not a bad idea to verify the ESR of the new parts you're installing

as well.

 -- john, KE5FX
 Miles Design LLC

 _______________________________________________
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

 and follow the instructions there.

time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

It looks rough, there's a cesium standard for sale in the UK, time nuts member perhaps, that's sat at a quarter of the price, I know which I'd buy. On 24 Feb 2018 10:46, "George Atkinson via time-nuts" <time-nuts@febo.com> wrote: > There is one on UK ebay at the moment but its not being given away. From > the one partial photograph it looks a bit rough. > > https://www.ebay.co.uk/itm/HP-5065A-Rubidium-Frequency- > Standard/183057846558? > > Robert G8RPI. > > > > > On 24 February 2018 at 10:02 John Miles <john@miles.io> wrote: > > > > > > > > > Hi > > > > > > Well, if anybody else is “giving away” 5065’s I’d certainly be > willing to > > > “accept” > > > one :) > > > > > > Bob > > > > > > > > > Be careful what you wish for. :) The unit Pete is talking about is a > 1969-era model, in nice overall shape but with the usual bad capacitors on > the oven controller board. The lamp oven winding should be about 50 ohms > cold, but is closer to a dead short. It will take a LOT of work to restore > to working condition. > > > > The failed controller ran long enough to smoke a 1.5-ohm 1W > resistor, 1N400-something diode, and eventually the 1A line fuse ( > http://www.ke5fx.com/5065A_A11_sm.jpg ). Needless to say the lamp PCB > looks like something out of Fukushima. It'll need to be rebuilt from > scratch after rewinding the heater. > > > > Seriously -- anyone with a 5065A who hasn't checked/replaced the > caps on A11, don't let this happen to you. They aren't making any more of > these puppies. If the caps on that board are original, don't bother to > check them, just replace them, as Luciano suggests at > http://www.timeok.it/hp5065a-corner-3/. No need for exotic parts, just > put in whatever you have that is somewhere close to the original values. > > > > Not a bad idea to verify the ESR of the new parts you're installing > as well. > > > > -- john, KE5FX > > Miles Design LLC > > > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > and follow the instructions there. >
BK
Bob kb8tq
Sat, Feb 24, 2018 2:30 PM

Hi

The “pickup only” part of the deal would be a bit of an issue for some of us :)

Bob

On Feb 24, 2018, at 5:46 AM, George Atkinson via time-nuts time-nuts@febo.com wrote:

There is one on UK ebay at the moment but its not being given away. From the one partial photograph it looks a bit rough.

https://www.ebay.co.uk/itm/HP-5065A-Rubidium-Frequency-Standard/183057846558?

Robert G8RPI.

On 24 February 2018 at 10:02 John Miles <john@miles.io> wrote:
    Hi

    Well, if anybody else is “giving away” 5065’s I’d certainly be willing to
    “accept”
    one :)

    Bob
Be careful what you wish for. :) The unit Pete is talking about is a 1969-era model, in nice overall shape but with the usual bad capacitors on the oven controller board. The lamp oven winding should be about 50 ohms cold, but is closer to a dead short. It will take a LOT of work to restore to working condition.

The failed controller ran long enough to smoke a 1.5-ohm 1W resistor, 1N400-something diode, and eventually the 1A line fuse ( http://www.ke5fx.com/5065A_A11_sm.jpg ). Needless to say the lamp PCB looks like something out of Fukushima. It'll need to be rebuilt from scratch after rewinding the heater.

Seriously -- anyone with a 5065A who hasn't checked/replaced the caps on A11, don't let this happen to you. They aren't making any more of these puppies. If the caps on that board are original, don't bother to check them, just replace them, as Luciano suggests at http://www.timeok.it/hp5065a-corner-3/. No need for exotic parts, just put in whatever you have that is somewhere close to the original values.

Not a bad idea to verify the ESR of the new parts you're installing as well.

-- john, KE5FX
Miles Design LLC

_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi The “pickup only” part of the deal would be a bit of an issue for some of us :) Bob > On Feb 24, 2018, at 5:46 AM, George Atkinson via time-nuts <time-nuts@febo.com> wrote: > > There is one on UK ebay at the moment but its not being given away. From the one partial photograph it looks a bit rough. > > https://www.ebay.co.uk/itm/HP-5065A-Rubidium-Frequency-Standard/183057846558? > > Robert G8RPI. > >> >> On 24 February 2018 at 10:02 John Miles <john@miles.io> wrote: >> >>>> >>> Hi >>> >>> Well, if anybody else is “giving away” 5065’s I’d certainly be willing to >>> “accept” >>> one :) >>> >>> Bob >>> >>>> >> Be careful what you wish for. :) The unit Pete is talking about is a 1969-era model, in nice overall shape but with the usual bad capacitors on the oven controller board. The lamp oven winding should be about 50 ohms cold, but is closer to a dead short. It will take a LOT of work to restore to working condition. >> >> The failed controller ran long enough to smoke a 1.5-ohm 1W resistor, 1N400-something diode, and eventually the 1A line fuse ( http://www.ke5fx.com/5065A_A11_sm.jpg ). Needless to say the lamp PCB looks like something out of Fukushima. It'll need to be rebuilt from scratch after rewinding the heater. >> >> Seriously -- anyone with a 5065A who hasn't checked/replaced the caps on A11, don't let this happen to you. They aren't making any more of these puppies. If the caps on that board are original, don't bother to check them, just replace them, as Luciano suggests at http://www.timeok.it/hp5065a-corner-3/. No need for exotic parts, just put in whatever you have that is somewhere close to the original values. >> >> Not a bad idea to verify the ESR of the new parts you're installing as well. >> >> -- john, KE5FX >> Miles Design LLC >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
E
ewkehren
Sat, Feb 24, 2018 2:49 PM

Old color and a little expensive only one partial pictureBert Kehren

Sent from my Galaxy Tab® A
-------- Original message --------From: Bob kb8tq kb8tq@n1k.org Date: 2/24/18  9:30 AM  (GMT-05:00) To: George Atkinson robertg8rpi@virginmedia.com, Discussion of precise time and frequency measurement time-nuts@febo.com Subject: Re: [time-nuts] Curious, how many t-nuts have 5065A
Hi

The “pickup only” part of the deal would be a bit of an issue for some of us :)

Bob

On Feb 24, 2018, at 5:46 AM, George Atkinson via time-nuts time-nuts@febo.com wrote:

There is one on UK ebay at the moment but its not being given away. From the one partial photograph it looks a bit rough.

https://www.ebay.co.uk/itm/HP-5065A-Rubidium-Frequency-Standard/183057846558?

Robert G8RPI.

    On 24 February 2018 at 10:02 John Miles john@miles.io wrote:

        Hi

        Well, if anybody else is “giving away” 5065’s I’d certainly be willing to
        “accept”
        one :)

        Bob

    Be careful what you wish for. :) The unit Pete is talking about is a 1969-era model, in nice overall shape but with the usual bad capacitors on the oven controller board. The lamp oven winding should be about 50 ohms cold, but is closer to a dead short. It will take a LOT of work to restore to working condition.

    The failed controller ran long enough to smoke a 1.5-ohm 1W resistor, 1N400-something diode, and eventually the 1A line fuse ( http://www.ke5fx.com/5065A_A11_sm.jpg ). Needless to say the lamp PCB looks like something out of Fukushima. It'll need to be rebuilt from scratch after rewinding the heater.

    Seriously -- anyone with a 5065A who hasn't checked/replaced the caps on A11, don't let this happen to you. They aren't making any more of these puppies. If the caps on that board are original, don't bother to check them, just replace them, as Luciano suggests at http://www.timeok.it/hp5065a-corner-3/. No need for exotic parts, just put in whatever you have that is somewhere close to the original values.

    Not a bad idea to verify the ESR of the new parts you're installing as well.

    -- john, KE5FX
    Miles Design LLC

    _______________________________________________
    time-nuts mailing list -- time-nuts@febo.com
    To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
    and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Old color and a little expensive only one partial pictureBert Kehren Sent from my Galaxy Tab® A -------- Original message --------From: Bob kb8tq <kb8tq@n1k.org> Date: 2/24/18 9:30 AM (GMT-05:00) To: George Atkinson <robertg8rpi@virginmedia.com>, Discussion of precise time and frequency measurement <time-nuts@febo.com> Subject: Re: [time-nuts] Curious, how many t-nuts have 5065A Hi The “pickup only” part of the deal would be a bit of an issue for some of us :) Bob > On Feb 24, 2018, at 5:46 AM, George Atkinson via time-nuts <time-nuts@febo.com> wrote: > > There is one on UK ebay at the moment but its not being given away. From the one partial photograph it looks a bit rough. > > https://www.ebay.co.uk/itm/HP-5065A-Rubidium-Frequency-Standard/183057846558? > > Robert G8RPI. > >> >>    On 24 February 2018 at 10:02 John Miles <john@miles.io> wrote: >> >>>> >>>        Hi >>> >>>        Well, if anybody else is “giving away” 5065’s I’d certainly be willing to >>>        “accept” >>>        one :) >>> >>>        Bob >>> >>>> >>    Be careful what you wish for. :) The unit Pete is talking about is a 1969-era model, in nice overall shape but with the usual bad capacitors on the oven controller board. The lamp oven winding should be about 50 ohms cold, but is closer to a dead short. It will take a LOT of work to restore to working condition. >> >>    The failed controller ran long enough to smoke a 1.5-ohm 1W resistor, 1N400-something diode, and eventually the 1A line fuse ( http://www.ke5fx.com/5065A_A11_sm.jpg ). Needless to say the lamp PCB looks like something out of Fukushima. It'll need to be rebuilt from scratch after rewinding the heater. >> >>    Seriously -- anyone with a 5065A who hasn't checked/replaced the caps on A11, don't let this happen to you. They aren't making any more of these puppies. If the caps on that board are original, don't bother to check them, just replace them, as Luciano suggests at http://www.timeok.it/hp5065a-corner-3/. No need for exotic parts, just put in whatever you have that is somewhere close to the original values. >> >>    Not a bad idea to verify the ESR of the new parts you're installing as well. >> >>    -- john, KE5FX >>    Miles Design LLC >> >>    _______________________________________________ >>    time-nuts mailing list -- time-nuts@febo.com >>    To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>    and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.