time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

Why is holdover LED on HP 58503A not lit, when GPS lock is unlit too?

DD
Dr. David Kirkby (Kirkby Microwave Ltd)
Sun, Dec 18, 2016 1:01 PM

My 58503 GPS time/frequency reference had its power disconnected a few
times yesterday as I moved things in the lab. At the modem the status of
the LEDs are

  • Power green
  • GPS lock - not lit
  • Holdover - not lit
  • Alarm - not lit.

That is what the manual says will happen  when power is first applied, but
I'm puzzled why.

The SYSTEM:STATUS? shows the information at the end of this email. What I
can't understand is why the holdover light is not lit, when clearly its not
tracking any satellites. It seems logical to me that holdover should be on,
when GPS lock is not, and visa versa. But that does not seem to be the
case.

Looking at the data below, the date/time is obviously wrong (01 Jan 1998),
the height at -14.04 m  (MSL) is wrong, but latitude and longitude look
about right, although they are certainly not exactly agreeing with Google
maps, which show 51°39'04.1"N+0°46'36.4"E.

I must be misunderstanding the purpose of these lights.

I'm also a bit puzzled it has been on about 1.5 hours, and can't seem to
find a single satellite. The antenna is fairly clear of anything else, and
it was certainly working yesterday, with the power and GPS lock LEDs both
lit.

Any ideas what's going on?

scpi > SYSTEM:STATUS?
------------------------------- Receiver Status

SYNCHRONIZATION ........................................... [ Outputs
Invalid ]
SmartClock Mode ___________________________  Reference Outputs


Locked                                    TFOM    9
FFOM    3
Recovery                                  1PPS TI  --
Holdover                                  HOLD THR 1.000 us

Power-up: GPS acquisition                  Holdover Uncertainty


                                          Predict  --

ACQUISITION .............................................. [ GPS 1PPS
Invalid ]
Tracking: 0 ____  Not Tracking: 1 ________  Time _____ +1 leap second
pending
PRN  El  Az                UTC      12:59:16 [?] 01 Jan
1998
*32  -- ---                GPS 1PPS Invalid: not tracking
ANT DLY  0 ns
Position


                                          MODE     Survey:      0%

complete
Suspended: track <4
sats
INIT LAT N  51:39:03.825
INIT LON E  0:46:36.278
ELEV MASK 10 deg  *attempting to track      INIT HGT          -14.04 m
(MSL)
HEALTH MONITOR ......................................................... [
OK ]
Self Test: OK    Int Pwr: OK  Oven Pwr: OK  OCXO: OK  EFC: OK  GPS Rcv:
OK
scpi >

Dr. David Kirkby Ph.D CEng MIET
Kirkby Microwave Ltd
Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, Essex, CM3 6DT,
UK.
Registered in England and Wales, company number 08914892.
http://www.kirkbymicrowave.co.uk/
Tel: 07910 441670 / +44 7910 441670 (0900 to 2100 GMT only please)

My 58503 GPS time/frequency reference had its power disconnected a few times yesterday as I moved things in the lab. At the modem the status of the LEDs are * Power green * GPS lock - not lit * Holdover - not lit * Alarm - not lit. That is what the manual says will happen when power is first applied, but I'm puzzled why. The SYSTEM:STATUS? shows the information at the end of this email. What I can't understand is why the holdover light is not lit, when clearly its not tracking any satellites. It seems logical to me that holdover should be on, when GPS lock is not, and visa versa. But that does not seem to be the case. Looking at the data below, the date/time is obviously wrong (01 Jan 1998), the height at -14.04 m (MSL) is wrong, but latitude and longitude look about right, although they are certainly not exactly agreeing with Google maps, which show 51°39'04.1"N+0°46'36.4"E. I must be misunderstanding the purpose of these lights. I'm also a bit puzzled it has been on about 1.5 hours, and can't seem to find a single satellite. The antenna is fairly clear of anything else, and it was certainly working yesterday, with the power and GPS lock LEDs both lit. Any ideas what's going on? scpi > SYSTEM:STATUS? ------------------------------- Receiver Status ------------------------------- SYNCHRONIZATION ........................................... [ Outputs Invalid ] SmartClock Mode ___________________________ Reference Outputs _______________ Locked TFOM 9 FFOM 3 Recovery 1PPS TI -- Holdover HOLD THR 1.000 us >> Power-up: GPS acquisition Holdover Uncertainty ____________ Predict -- ACQUISITION .............................................. [ GPS 1PPS Invalid ] Tracking: 0 ____ Not Tracking: 1 ________ Time _____ +1 leap second pending PRN El Az UTC 12:59:16 [?] 01 Jan 1998 *32 -- --- GPS 1PPS Invalid: not tracking ANT DLY 0 ns Position ________________________ MODE Survey: 0% complete Suspended: track <4 sats INIT LAT N 51:39:03.825 INIT LON E 0:46:36.278 ELEV MASK 10 deg *attempting to track INIT HGT -14.04 m (MSL) HEALTH MONITOR ......................................................... [ OK ] Self Test: OK Int Pwr: OK Oven Pwr: OK OCXO: OK EFC: OK GPS Rcv: OK scpi > Dr. David Kirkby Ph.D CEng MIET Kirkby Microwave Ltd Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, Essex, CM3 6DT, UK. Registered in England and Wales, company number 08914892. http://www.kirkbymicrowave.co.uk/ Tel: 07910 441670 / +44 7910 441670 (0900 to 2100 GMT only please)
AM
Artek Manuals
Sun, Dec 18, 2016 2:27 PM

I believe  the "Holdover" function only occurs when you you loose just
the  satellites

Once locked disconnect the antenna for a minute and you will see the
holdover LED light.

IF the unit looses POWER it has no way of of powering the oscillator and
its ovens ( to do a holdover) and has no way of knowing how long the
unit has been off when re-powered until it goes through a complete
acquisition/re-sync cycle.

DAVE

manuals@artekmanuals.com

On 12/18/2016 8:01 AM, Dr. David Kirkby (Kirkby Microwave Ltd) wrote:

My 58503 GPS time/frequency reference had its power disconnected a few
times yesterday as I moved things in the lab. At the modem the status of
the LEDs are

  • Power green
  • GPS lock - not lit
  • Holdover - not lit
  • Alarm - not lit.

That is what the manual says will happen  when power is first applied, but
I'm puzzled why.

The SYSTEM:STATUS? shows the information at the end of this email. What I
can't understand is why the holdover light is not lit, when clearly its not
tracking any satellites. It seems logical to me that holdover should be on,
when GPS lock is not, and visa versa. But that does not seem to be the
case.

Looking at the data below, the date/time is obviously wrong (01 Jan 1998),
the height at -14.04 m  (MSL) is wrong, but latitude and longitude look
about right, although they are certainly not exactly agreeing with Google
maps, which show 51°39'04.1"N+0°46'36.4"E.

I must be misunderstanding the purpose of these lights.

I'm also a bit puzzled it has been on about 1.5 hours, and can't seem to
find a single satellite. The antenna is fairly clear of anything else, and
it was certainly working yesterday, with the power and GPS lock LEDs both
lit.

Any ideas what's going on?

scpi > SYSTEM:STATUS?
------------------------------- Receiver Status

SYNCHRONIZATION ........................................... [ Outputs
Invalid ]
SmartClock Mode ___________________________  Reference Outputs


 Locked                                     TFOM     9

FFOM    3
Recovery                                  1PPS TI  --
Holdover                                  HOLD THR 1.000 us

Power-up: GPS acquisition                  Holdover Uncertainty


                                            Predict  --

ACQUISITION .............................................. [ GPS 1PPS
Invalid ]
Tracking: 0 ____  Not Tracking: 1 ________  Time _____ +1 leap second
pending
PRN  El  Az                UTC      12:59:16 [?] 01 Jan
1998
*32  -- ---                GPS 1PPS Invalid: not tracking
ANT DLY  0 ns
Position


                                            MODE     Survey:      0%

complete
Suspended: track <4
sats
INIT LAT N  51:39:03.825
INIT LON E  0:46:36.278
ELEV MASK 10 deg  *attempting to track      INIT HGT          -14.04 m
(MSL)
HEALTH MONITOR ......................................................... [
OK ]
Self Test: OK    Int Pwr: OK  Oven Pwr: OK  OCXO: OK  EFC: OK  GPS Rcv:
OK
scpi >

Dr. David Kirkby Ph.D CEng MIET
Kirkby Microwave Ltd
Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, Essex, CM3 6DT,
UK.
Registered in England and Wales, company number 08914892.
http://www.kirkbymicrowave.co.uk/
Tel: 07910 441670 / +44 7910 441670 (0900 to 2100 GMT only please)


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

I believe the "Holdover" function only occurs when you you loose just the satellites Once locked disconnect the antenna for a minute and you will see the holdover LED light. IF the unit looses POWER it has no way of of powering the oscillator and its ovens ( to do a holdover) and has no way of knowing how long the unit has been off when re-powered until it goes through a complete acquisition/re-sync cycle. DAVE manuals@artekmanuals.com On 12/18/2016 8:01 AM, Dr. David Kirkby (Kirkby Microwave Ltd) wrote: > My 58503 GPS time/frequency reference had its power disconnected a few > times yesterday as I moved things in the lab. At the modem the status of > the LEDs are > > * Power green > * GPS lock - not lit > * Holdover - not lit > * Alarm - not lit. > > That is what the manual says will happen when power is first applied, but > I'm puzzled why. > > The SYSTEM:STATUS? shows the information at the end of this email. What I > can't understand is why the holdover light is not lit, when clearly its not > tracking any satellites. It seems logical to me that holdover should be on, > when GPS lock is not, and visa versa. But that does not seem to be the > case. > > Looking at the data below, the date/time is obviously wrong (01 Jan 1998), > the height at -14.04 m (MSL) is wrong, but latitude and longitude look > about right, although they are certainly not exactly agreeing with Google > maps, which show 51°39'04.1"N+0°46'36.4"E. > > I must be misunderstanding the purpose of these lights. > > I'm also a bit puzzled it has been on about 1.5 hours, and can't seem to > find a single satellite. The antenna is fairly clear of anything else, and > it was certainly working yesterday, with the power and GPS lock LEDs both > lit. > > Any ideas what's going on? > > scpi > SYSTEM:STATUS? > ------------------------------- Receiver Status > ------------------------------- > > SYNCHRONIZATION ........................................... [ Outputs > Invalid ] > SmartClock Mode ___________________________ Reference Outputs > _______________ > Locked TFOM 9 > FFOM 3 > Recovery 1PPS TI -- > Holdover HOLD THR 1.000 us >>> Power-up: GPS acquisition Holdover Uncertainty > ____________ > Predict -- > > ACQUISITION .............................................. [ GPS 1PPS > Invalid ] > Tracking: 0 ____ Not Tracking: 1 ________ Time _____ +1 leap second > pending > PRN El Az UTC 12:59:16 [?] 01 Jan > 1998 > *32 -- --- GPS 1PPS Invalid: not tracking > ANT DLY 0 ns > Position > ________________________ > MODE Survey: 0% > complete > Suspended: track <4 > sats > INIT LAT N 51:39:03.825 > INIT LON E 0:46:36.278 > ELEV MASK 10 deg *attempting to track INIT HGT -14.04 m > (MSL) > HEALTH MONITOR ......................................................... [ > OK ] > Self Test: OK Int Pwr: OK Oven Pwr: OK OCXO: OK EFC: OK GPS Rcv: > OK > scpi > > > > > Dr. David Kirkby Ph.D CEng MIET > Kirkby Microwave Ltd > Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, Essex, CM3 6DT, > UK. > Registered in England and Wales, company number 08914892. > http://www.kirkbymicrowave.co.uk/ > Tel: 07910 441670 / +44 7910 441670 (0900 to 2100 GMT only please) > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > -- Dave Manuals@ArtekManuals.com www.ArtekManuals.com
TS
Tim Shoppa
Sun, Dec 18, 2016 2:31 PM

A common misconception, is that holdover is the opposite of GPS lock.
Sometimes we might even talk about the two as if we're in one or the other.
But really the power-on state, is that we're in neither holdover or GPS
lock.

Holdover means the smartclock previously had GPS lock and had used it to
characterize the aging of the OCXO, and is maintaining the EFC
extrapolation, such that it believes it can still produce accurate time and
frequency through the GPS loss.

The state of your clock at the moment, is that it has never had GPS lock
since regaining power, it does not know what time it is, and it has not
characterized or extrapolated the aging of the OCXO. So it is definitely
not in holdover.

Tim N3QE

On Sun, Dec 18, 2016 at 8:01 AM, Dr. David Kirkby (Kirkby Microwave Ltd) <
drkirkby@kirkbymicrowave.co.uk> wrote:

My 58503 GPS time/frequency reference had its power disconnected a few
times yesterday as I moved things in the lab. At the modem the status of
the LEDs are

  • Power green
  • GPS lock - not lit
  • Holdover - not lit
  • Alarm - not lit.

That is what the manual says will happen  when power is first applied, but
I'm puzzled why.

The SYSTEM:STATUS? shows the information at the end of this email. What I
can't understand is why the holdover light is not lit, when clearly its not
tracking any satellites. It seems logical to me that holdover should be on,
when GPS lock is not, and visa versa. But that does not seem to be the
case.

Looking at the data below, the date/time is obviously wrong (01 Jan 1998),
the height at -14.04 m  (MSL) is wrong, but latitude and longitude look
about right, although they are certainly not exactly agreeing with Google
maps, which show 51°39'04.1"N+0°46'36.4"E.

I must be misunderstanding the purpose of these lights.

I'm also a bit puzzled it has been on about 1.5 hours, and can't seem to
find a single satellite. The antenna is fairly clear of anything else, and
it was certainly working yesterday, with the power and GPS lock LEDs both
lit.

Any ideas what's going on?

scpi > SYSTEM:STATUS?
------------------------------- Receiver Status

SYNCHRONIZATION ........................................... [ Outputs
Invalid ]
SmartClock Mode ___________________________  Reference Outputs


Locked                                     TFOM     9

FFOM    3
Recovery                                  1PPS TI  --
Holdover                                  HOLD THR 1.000 us

Power-up: GPS acquisition                  Holdover Uncertainty


                                           Predict  --

ACQUISITION .............................................. [ GPS 1PPS
Invalid ]
Tracking: 0 ____  Not Tracking: 1 ________  Time _____ +1 leap second
pending
PRN  El  Az                UTC      12:59:16 [?] 01 Jan
1998
*32  -- ---                GPS 1PPS Invalid: not
tracking
ANT DLY  0 ns
Position


                                           MODE     Survey:      0%

complete
Suspended: track <4
sats
INIT LAT N  51:39:03.825
INIT LON E  0:46:36.278
ELEV MASK 10 deg  *attempting to track      INIT HGT          -14.04 m
(MSL)
HEALTH MONITOR ......................................................... [
OK ]
Self Test: OK    Int Pwr: OK  Oven Pwr: OK  OCXO: OK  EFC: OK  GPS Rcv:
OK
scpi >

Dr. David Kirkby Ph.D CEng MIET
Kirkby Microwave Ltd
Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, Essex, CM3 6DT,
UK.
Registered in England and Wales, company number 08914892.
http://www.kirkbymicrowave.co.uk/
Tel: 07910 441670 / +44 7910 441670 (0900 to 2100 GMT only please)


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

A common misconception, is that holdover is the opposite of GPS lock. Sometimes we might even talk about the two as if we're in one or the other. But really the power-on state, is that we're in neither holdover or GPS lock. Holdover means the smartclock previously had GPS lock and had used it to characterize the aging of the OCXO, and is maintaining the EFC extrapolation, such that it believes it can still produce accurate time and frequency through the GPS loss. The state of your clock at the moment, is that it has never had GPS lock since regaining power, it does not know what time it is, and it has not characterized or extrapolated the aging of the OCXO. So it is definitely not in holdover. Tim N3QE On Sun, Dec 18, 2016 at 8:01 AM, Dr. David Kirkby (Kirkby Microwave Ltd) < drkirkby@kirkbymicrowave.co.uk> wrote: > My 58503 GPS time/frequency reference had its power disconnected a few > times yesterday as I moved things in the lab. At the modem the status of > the LEDs are > > * Power green > * GPS lock - not lit > * Holdover - not lit > * Alarm - not lit. > > That is what the manual says will happen when power is first applied, but > I'm puzzled why. > > The SYSTEM:STATUS? shows the information at the end of this email. What I > can't understand is why the holdover light is not lit, when clearly its not > tracking any satellites. It seems logical to me that holdover should be on, > when GPS lock is not, and visa versa. But that does not seem to be the > case. > > Looking at the data below, the date/time is obviously wrong (01 Jan 1998), > the height at -14.04 m (MSL) is wrong, but latitude and longitude look > about right, although they are certainly not exactly agreeing with Google > maps, which show 51°39'04.1"N+0°46'36.4"E. > > I must be misunderstanding the purpose of these lights. > > I'm also a bit puzzled it has been on about 1.5 hours, and can't seem to > find a single satellite. The antenna is fairly clear of anything else, and > it was certainly working yesterday, with the power and GPS lock LEDs both > lit. > > Any ideas what's going on? > > scpi > SYSTEM:STATUS? > ------------------------------- Receiver Status > ------------------------------- > > SYNCHRONIZATION ........................................... [ Outputs > Invalid ] > SmartClock Mode ___________________________ Reference Outputs > _______________ > Locked TFOM 9 > FFOM 3 > Recovery 1PPS TI -- > Holdover HOLD THR 1.000 us > >> Power-up: GPS acquisition Holdover Uncertainty > ____________ > Predict -- > > ACQUISITION .............................................. [ GPS 1PPS > Invalid ] > Tracking: 0 ____ Not Tracking: 1 ________ Time _____ +1 leap second > pending > PRN El Az UTC 12:59:16 [?] 01 Jan > 1998 > *32 -- --- GPS 1PPS Invalid: not > tracking > ANT DLY 0 ns > Position > ________________________ > MODE Survey: 0% > complete > Suspended: track <4 > sats > INIT LAT N 51:39:03.825 > INIT LON E 0:46:36.278 > ELEV MASK 10 deg *attempting to track INIT HGT -14.04 m > (MSL) > HEALTH MONITOR ......................................................... [ > OK ] > Self Test: OK Int Pwr: OK Oven Pwr: OK OCXO: OK EFC: OK GPS Rcv: > OK > scpi > > > > > Dr. David Kirkby Ph.D CEng MIET > Kirkby Microwave Ltd > Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, Essex, CM3 6DT, > UK. > Registered in England and Wales, company number 08914892. > http://www.kirkbymicrowave.co.uk/ > Tel: 07910 441670 / +44 7910 441670 (0900 to 2100 GMT only please) > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > and follow the instructions there. >
DD
Dr. David Kirkby (Kirkby Microwave Ltd)
Sun, Dec 18, 2016 5:34 PM

On 18 December 2016 at 14:31, Tim Shoppa tshoppa@gmail.com wrote:

A common misconception, is that holdover is the opposite of GPS lock.
Sometimes we might even talk about the two as if we're in one or the other.
But really the power-on state, is that we're in neither holdover or GPS
lock.

Holdover means the smartclock previously had GPS lock and had used it to
characterize the aging of the OCXO, and is maintaining the EFC
extrapolation, such that it believes it can still produce accurate time and
frequency through the GPS loss.

The state of your clock at the moment, is that it has never had GPS lock
since regaining power, it does not know what time it is, and it has not
characterized or extrapolated the aging of the OCXO. So it is definitely
not in holdover.

Tim N3QE

Cheers Tim, that makes sense.

I'm a bit concerned it has not managed to track a single satellite, despite
it has been on 5 hours or so. But perhaps things have improved, as its not
attempting to track 6 (PRN=15, 2, 21, 25, 26 and 28), whereas before it was
attempting to track just with a PRN of 32. I'm hoping its current status
(see below), is looking more hopeful than what it was before.

scpi >
scpi > SYSTEM:STATUS?
------------------------------- Receiver Status

SYNCHRONIZATION ........................................... [ Outputs
Invalid ]
SmartClock Mode ___________________________  Reference Outputs


Locked                                    TFOM    9
FFOM    3
Recovery                                  1PPS TI  --
Holdover                                  HOLD THR 1.000 us

Power-up: GPS acquisition                  Holdover Uncertainty


                                          Predict  --

ACQUISITION .............................................. [ GPS 1PPS
Invalid ]
Tracking: 0 ____  Not Tracking: 6 ________  Time _____ +1 leap second
pending
PRN  El  Az                UTC      14:43:22 [?] 01 Jan
1998
*15  -- ---                GPS 1PPS Invalid: not tracking
*20  -- ---                ANT DLY  0 ns
*21  -- ---                Position


               *25  -- ---                MODE     Survey:      0%

complete
*26  -- ---                        Suspended: track <4
sats
*28  -- ---                INIT LAT N  51:39:03.825
INIT LON E  0:46:36.278
ELEV MASK 10 deg  *attempting to track      INIT HGT          -14.04 m
(MSL)
HEALTH MONITOR ......................................................... [
OK ]
Self Test: OK    Int Pwr: OK  Oven Pwr: OK  OCXO: OK  EFC: OK  GPS Rcv:
OK
scpi > SYSTEM:STATUS?
------------------------------- Receiver Status

SYNCHRONIZATION ........................................... [ Outputs
Invalid ]
SmartClock Mode ___________________________  Reference Outputs


Locked                                    TFOM    9
FFOM    3
Recovery                                  1PPS TI  --
Holdover                                  HOLD THR 1.000 us

Power-up: GPS acquisition                  Holdover Uncertainty


                                          Predict  --

ACQUISITION .............................................. [ GPS 1PPS
Invalid ]
Tracking: 0 ____  Not Tracking: 6 ________  Time _____ +1 leap second
pending
PRN  El  Az                UTC      16:59:08 [?] 01 Jan
1998
* 8  -- ---                GPS 1PPS Invalid: not tracking
*10  -- ---                ANT DLY  0 ns
*11  -- ---                Position


               *12  -- ---                MODE     Survey:      0%

complete
*13  -- ---                        Suspended: track <4
sats
*30  -- ---                INIT LAT N  51:39:03.825
INIT LON E  0:46:36.278
ELEV MASK 10 deg  *attempting to track      INIT HGT          -14.04 m
(MSL)
HEALTH MONITOR ......................................................... [
OK ]
Self Test: OK    Int Pwr: OK  Oven Pwr: OK  OCXO: OK  EFC: OK  GPS Rcv:
OK

On 18 December 2016 at 14:31, Tim Shoppa <tshoppa@gmail.com> wrote: > A common misconception, is that holdover is the opposite of GPS lock. > Sometimes we might even talk about the two as if we're in one or the other. > But really the power-on state, is that we're in neither holdover or GPS > lock. > > Holdover means the smartclock previously had GPS lock and had used it to > characterize the aging of the OCXO, and is maintaining the EFC > extrapolation, such that it believes it can still produce accurate time and > frequency through the GPS loss. > > The state of your clock at the moment, is that it has never had GPS lock > since regaining power, it does not know what time it is, and it has not > characterized or extrapolated the aging of the OCXO. So it is definitely > not in holdover. > > Tim N3QE > Cheers Tim, that makes sense. I'm a bit concerned it has not managed to track a single satellite, despite it has been on 5 hours or so. But perhaps things have improved, as its not attempting to track 6 (PRN=15, 2, 21, 25, 26 and 28), whereas before it was attempting to track just with a PRN of 32. I'm hoping its current status (see below), is looking more hopeful than what it was before. scpi > scpi > SYSTEM:STATUS? ------------------------------- Receiver Status ------------------------------- SYNCHRONIZATION ........................................... [ Outputs Invalid ] SmartClock Mode ___________________________ Reference Outputs _______________ Locked TFOM 9 FFOM 3 Recovery 1PPS TI -- Holdover HOLD THR 1.000 us >> Power-up: GPS acquisition Holdover Uncertainty ____________ Predict -- ACQUISITION .............................................. [ GPS 1PPS Invalid ] Tracking: 0 ____ Not Tracking: 6 ________ Time _____ +1 leap second pending PRN El Az UTC 14:43:22 [?] 01 Jan 1998 *15 -- --- GPS 1PPS Invalid: not tracking *20 -- --- ANT DLY 0 ns *21 -- --- Position ________________________ *25 -- --- MODE Survey: 0% complete *26 -- --- Suspended: track <4 sats *28 -- --- INIT LAT N 51:39:03.825 INIT LON E 0:46:36.278 ELEV MASK 10 deg *attempting to track INIT HGT -14.04 m (MSL) HEALTH MONITOR ......................................................... [ OK ] Self Test: OK Int Pwr: OK Oven Pwr: OK OCXO: OK EFC: OK GPS Rcv: OK scpi > SYSTEM:STATUS? ------------------------------- Receiver Status ------------------------------- SYNCHRONIZATION ........................................... [ Outputs Invalid ] SmartClock Mode ___________________________ Reference Outputs _______________ Locked TFOM 9 FFOM 3 Recovery 1PPS TI -- Holdover HOLD THR 1.000 us >> Power-up: GPS acquisition Holdover Uncertainty ____________ Predict -- ACQUISITION .............................................. [ GPS 1PPS Invalid ] Tracking: 0 ____ Not Tracking: 6 ________ Time _____ +1 leap second pending PRN El Az UTC 16:59:08 [?] 01 Jan 1998 * 8 -- --- GPS 1PPS Invalid: not tracking *10 -- --- ANT DLY 0 ns *11 -- --- Position ________________________ *12 -- --- MODE Survey: 0% complete *13 -- --- Suspended: track <4 sats *30 -- --- INIT LAT N 51:39:03.825 INIT LON E 0:46:36.278 ELEV MASK 10 deg *attempting to track INIT HGT -14.04 m (MSL) HEALTH MONITOR ......................................................... [ OK ] Self Test: OK Int Pwr: OK Oven Pwr: OK OCXO: OK EFC: OK GPS Rcv: OK
DD
Dr. David Kirkby (Kirkby Microwave Ltd)
Sun, Dec 18, 2016 5:39 PM

On 18 December 2016 at 17:34, Dr. David Kirkby (Kirkby Microwave Ltd) <
drkirkby@kirkbymicrowave.co.uk> wrote:

But perhaps things have improved, as its not attempting to track 6
(PRN=15, 2, 21, 25, 26 and 28), whereas before it was attempting to track
just with a PRN of 32. I'm hoping its current status (see below), is
looking more hopeful than what it was before.

I mean it is NOW attempting to track 6 !!! Almost the opposite of what I
wrote.

I'm guessing it gave up attempting to track the satellite with a PRN of 32.

Dave

On 18 December 2016 at 17:34, Dr. David Kirkby (Kirkby Microwave Ltd) < drkirkby@kirkbymicrowave.co.uk> wrote: > But perhaps things have improved, as its not attempting to track 6 > (PRN=15, 2, 21, 25, 26 and 28), whereas before it was attempting to track > just with a PRN of 32. I'm hoping its current status (see below), is > looking more hopeful than what it was before. > I mean it is NOW attempting to track 6 !!! Almost the opposite of what I wrote. I'm guessing it gave up attempting to track the satellite with a PRN of 32. Dave
DR
Dan Rae
Sun, Dec 18, 2016 6:19 PM

On 12/18/2016 9:34 AM, Dr. David Kirkby (Kirkby Microwave Ltd) wrote:


                 *12  -- ---                MODE     Survey:      0%

complete

If it's still showing that after being on for five hours, maybe it needs
to be explicitly told to start a survey?

dr

On 12/18/2016 9:34 AM, Dr. David Kirkby (Kirkby Microwave Ltd) wrote: > ________________________ > *12 -- --- MODE Survey: 0% > complete > If it's still showing that after being on for five hours, maybe it needs to be explicitly told to start a survey? dr
DD
Dr. David Kirkby (Kirkby Microwave Ltd)
Sun, Dec 18, 2016 10:14 PM

On 18 December 2016 at 18:19, Dan Rae danrae@verizon.net wrote:

On 12/18/2016 9:34 AM, Dr. David Kirkby (Kirkby Microwave Ltd) wrote:


                 *12  -- ---                MODE     Survey:      0%

complete

If it's still showing that after being on for five hours, maybe it needs
to be explicitly told to start a survey?

dr

I seem to be having some issues now with this. I've reset it, and entered
the approximate latitude, longitude, height, as well as date and time. The
commands below were executed, although NOT necessarily in the order give.

E-113> GPS:INIT:DATe 2016,12,18
E-113> gps:init:time 19,18,35
E-113> SYSTEM:PRESET
scpi > DIAG:LOG:READ:ALL?
Log status: 2  entries

Log 001:19970504.00:00:00:  Log cleared
Log 002:19970504.00:00:00:  System preset

The receiver seems to find satellites fairly quickly, tries to track them,
and is showing reasonable values for the things I set, like date, time,
position etc. However,  the receiver doesn't seem able to track any
satellites, so not surprisingly the survey never starts. I assume it wont
start until it has tracked sufficient number of satellites.

scpi > SYSTEM:STATUS?
------------------------------- Receiver Status

SYNCHRONIZATION ........................................... [ Outputs
Invalid ]
SmartClock Mode ___________________________  Reference Outputs


Locked                                    TFOM    9
FFOM    3
Recovery                                  1PPS TI  --
Holdover                                  HOLD THR 1.000 us

Power-up: GPS acquisition                  Holdover Uncertainty


                                          Predict  --

ACQUISITION .............................................. [ GPS 1PPS
Invalid ]
Tracking: 0 ____  Not Tracking: 12 _______  Time


               PRN  El  Az  PRN  El  Az   UTC      21:57:41 [?] 18 Dec

2016
6  20  67  12  76  56  GPS 1PPS Invalid: not tracking
* 7  -- ---  14  32 307  ANT DLY  0 ns
* 8  -- ---  24  42 129  Position


               * 9  -- ---   25  58 259   MODE     Survey:      0%

complete
*10  -- ---  *31  10 299            Suspended: track <4
sats
*11  -- ---  32  45 274  INIT LAT N  51:39:03.825
INIT LON E  0:46:36.277
ELEV MASK 10 deg  *attempting to track      INIT HGT          +29.99 m
(MSL)
HEALTH MONITOR ......................................................... [
OK ]
Self Test: OK    Int Pwr: OK  Oven Pwr: OK  OCXO: OK  EFC: OK  GPS Rcv:
OK

It will run overnight. Perhaps tomorrow it will sort itself out, but I'm
not too hopeful.

Dr. David Kirkby Ph.D CEng MIET
Kirkby Microwave Ltd
Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, Essex, CM3 6DT,
UK.
Registered in England and Wales, company number 08914892.
http://www.kirkbymicrowave.co.uk/
Tel: 07910 441670 / +44 7910 441670 (0900 to 2100 GMT only please)

On 18 December 2016 at 18:19, Dan Rae <danrae@verizon.net> wrote: > On 12/18/2016 9:34 AM, Dr. David Kirkby (Kirkby Microwave Ltd) wrote: > >> ________________________ >> *12 -- --- MODE Survey: 0% >> complete >> >> > If it's still showing that after being on for five hours, maybe it needs > to be explicitly told to start a survey? > > dr > I seem to be having some issues now with this. I've reset it, and entered the approximate latitude, longitude, height, as well as date and time. The commands below were executed, although NOT necessarily in the order give. E-113> GPS:INIT:DATe 2016,12,18 E-113> gps:init:time 19,18,35 E-113> SYSTEM:PRESET scpi > DIAG:LOG:READ:ALL? Log status: 2 entries Log 001:19970504.00:00:00: Log cleared Log 002:19970504.00:00:00: System preset The receiver seems to find satellites fairly quickly, tries to track them, and is showing reasonable values for the things I set, like date, time, position etc. However, the receiver doesn't seem able to track any satellites, so not surprisingly the survey never starts. I assume it wont start until it has tracked sufficient number of satellites. scpi > SYSTEM:STATUS? ------------------------------- Receiver Status ------------------------------- SYNCHRONIZATION ........................................... [ Outputs Invalid ] SmartClock Mode ___________________________ Reference Outputs _______________ Locked TFOM 9 FFOM 3 Recovery 1PPS TI -- Holdover HOLD THR 1.000 us >> Power-up: GPS acquisition Holdover Uncertainty ____________ Predict -- ACQUISITION .............................................. [ GPS 1PPS Invalid ] Tracking: 0 ____ Not Tracking: 12 _______ Time ____________________________ PRN El Az PRN El Az UTC 21:57:41 [?] 18 Dec 2016 6 20 67 12 76 56 GPS 1PPS Invalid: not tracking * 7 -- --- 14 32 307 ANT DLY 0 ns * 8 -- --- 24 42 129 Position ________________________ * 9 -- --- 25 58 259 MODE Survey: 0% complete *10 -- --- *31 10 299 Suspended: track <4 sats *11 -- --- 32 45 274 INIT LAT N 51:39:03.825 INIT LON E 0:46:36.277 ELEV MASK 10 deg *attempting to track INIT HGT +29.99 m (MSL) HEALTH MONITOR ......................................................... [ OK ] Self Test: OK Int Pwr: OK Oven Pwr: OK OCXO: OK EFC: OK GPS Rcv: OK It will run overnight. Perhaps tomorrow it will sort itself out, but I'm not too hopeful. Dr. David Kirkby Ph.D CEng MIET Kirkby Microwave Ltd Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, Essex, CM3 6DT, UK. Registered in England and Wales, company number 08914892. http://www.kirkbymicrowave.co.uk/ Tel: 07910 441670 / +44 7910 441670 (0900 to 2100 GMT only please)
DD
Dr. David Kirkby (Kirkby Microwave Ltd)
Sun, Dec 18, 2016 10:41 PM

It is looking much better now. I

  1. Powered off
  2. Left off for 30 s
  3. Pulled antenna out.
  4. Powered on
  5. Connected antenna, making sure if was firmly screwed in.

It then got GPS lock in a few minutes. It is running at reduced accuracy,
with the survey only 9.7% complete, but it is looking more hopeful.

I can't help feeling there should have been more informative status
available by the LEDs. The fact it is locked, but at reduced accuracy,
should really be visable from the front panel IMHO. Anyway, it seems to be
getting there - see below.

scpi >  SYSTEM:STATUS?
------------------------------- Receiver Status

SYNCHRONIZATION ............................ [ Outputs Valid/Reduced
Accuracy ]
SmartClock Mode ___________________________  Reference Outputs


Locked to GPS: stabilizing frequency      TFOM    3

FFOM    1
Recovery                                  1PPS TI -15.9 ns relative to
GPS
Holdover                                  HOLD THR 1.000 us
Power-up                                  Holdover Uncertainty


                                          Predict  432.0 us/initial 24

hrs

ACQUISITION ................................................ [ GPS 1PPS
Valid ]
Tracking: 6 ____  Not Tracking: 3 ________  Time _____ +1 leap second
pending
PRN  El  Az  SS  PRN  El  Az                UTC      22:39:23    18 Dec
2016
2  28  87  86    6  20  48                GPS 1PPS Synchronized to UTC
12  57  70  119    24  24 137                ANT DLY  0 ns
14  39 287  106    31  27 303                Position


25  76 284  96                              MODE    Survey:    9.7%
complete
29  37 194  79
32  39 249  109                              AVG LAT  N  51:39:04.205
AVG LON  E  0:46:36.347
ELEV MASK 10 deg                              AVG HGT          +42.42 m
(MSL)
HEALTH MONITOR ......................................................... [
OK ]
Self Test: OK    Int Pwr: OK  Oven Pwr: OK  OCXO: OK  EFC: OK  GPS Rcv:
OK
scpi >

It is looking much better now. I 1) Powered off 2) Left off for 30 s 3) Pulled antenna out. 4) Powered on 5) Connected antenna, making sure if was firmly screwed in. It then got GPS lock in a few minutes. It is running at reduced accuracy, with the survey only 9.7% complete, but it is looking more hopeful. I can't help feeling there should have been more informative status available by the LEDs. The fact it is locked, but at reduced accuracy, should really be visable from the front panel IMHO. Anyway, it seems to be getting there - see below. scpi > SYSTEM:STATUS? ------------------------------- Receiver Status ------------------------------- SYNCHRONIZATION ............................ [ Outputs Valid/Reduced Accuracy ] SmartClock Mode ___________________________ Reference Outputs _______________ >> Locked to GPS: stabilizing frequency TFOM 3 FFOM 1 Recovery 1PPS TI -15.9 ns relative to GPS Holdover HOLD THR 1.000 us Power-up Holdover Uncertainty ____________ Predict 432.0 us/initial 24 hrs ACQUISITION ................................................ [ GPS 1PPS Valid ] Tracking: 6 ____ Not Tracking: 3 ________ Time _____ +1 leap second pending PRN El Az SS PRN El Az UTC 22:39:23 18 Dec 2016 2 28 87 86 6 20 48 GPS 1PPS Synchronized to UTC 12 57 70 119 24 24 137 ANT DLY 0 ns 14 39 287 106 31 27 303 Position ________________________ 25 76 284 96 MODE Survey: 9.7% complete 29 37 194 79 32 39 249 109 AVG LAT N 51:39:04.205 AVG LON E 0:46:36.347 ELEV MASK 10 deg AVG HGT +42.42 m (MSL) HEALTH MONITOR ......................................................... [ OK ] Self Test: OK Int Pwr: OK Oven Pwr: OK OCXO: OK EFC: OK GPS Rcv: OK scpi >
DD
Dr. David Kirkby (Kirkby Microwave Ltd)
Mon, Dec 19, 2016 11:15 PM

On 18 December 2016 at 22:41, Dr. David Kirkby (Kirkby Microwave Ltd) <
drkirkby@kirkbymicrowave.co.uk> wrote:

It is looking much better now. I

  1. Powered off
  2. Left off for 30 s
  3. Pulled antenna out.
  4. Powered on
  5. Connected antenna, making sure if was firmly screwed in.

It then got GPS lock in a few minutes. It is running at reduced accuracy,
with the survey only 9.7% complete, but it is looking more hopeful.

This does appear now to be working - see below. I'm not exactly sure what
was the secret to getting this to work, but the above procedure seemed ok.

scpi >  SYSTEM:STATUS?
------------------------------- Receiver Status

SYNCHRONIZATION ............................................. [ Outputs
Valid ]
SmartClock Mode ___________________________  Reference Outputs


Locked to GPS                              TFOM    3

FFOM    0
Recovery                                  1PPS TI -8.3 ns relative to
GPS
Holdover                                  HOLD THR 1.000 us
Power-up                                  Holdover Uncertainty


                                          Predict  9.6 us/initial 24 hrs

ACQUISITION ................................................ [ GPS 1PPS
Valid ]
Tracking: 6 ____  Not Tracking: 2 ________  Time _____ +1 leap second
pending
PRN  El  Az  SS  PRN  El  Az                UTC      23:11:07    19 Dec
2016
2  31  69  83    6  15  34                GPS 1PPS Synchronized to UTC
12  41  80  110    32  27 234                ANT DLY  0 ns
14  37 266  81                              Position


25  83  33  127                              MODE    Hold
29  55 195  135
31  42 301  75                              LAT      N  51:39:04.198
LON      E  0:46:36.358
ELEV MASK 10 deg                              HGT              +45.42 m
(MSL)
HEALTH MONITOR ......................................................... [
OK ]
Self Test: OK    Int Pwr: OK  Oven Pwr: OK  OCXO: OK  EFC: OK  GPS Rcv:
OK
scpi >

On 18 December 2016 at 22:41, Dr. David Kirkby (Kirkby Microwave Ltd) < drkirkby@kirkbymicrowave.co.uk> wrote: > It is looking much better now. I > > 1) Powered off > 2) Left off for 30 s > 3) Pulled antenna out. > 4) Powered on > 5) Connected antenna, making sure if was firmly screwed in. > > It then got GPS lock in a few minutes. It is running at reduced accuracy, > with the survey only 9.7% complete, but it is looking more hopeful. > This does appear now to be working - see below. I'm not exactly sure what was the secret to getting this to work, but the above procedure seemed ok. scpi > SYSTEM:STATUS? ------------------------------- Receiver Status ------------------------------- SYNCHRONIZATION ............................................. [ Outputs Valid ] SmartClock Mode ___________________________ Reference Outputs _______________ >> Locked to GPS TFOM 3 FFOM 0 Recovery 1PPS TI -8.3 ns relative to GPS Holdover HOLD THR 1.000 us Power-up Holdover Uncertainty ____________ Predict 9.6 us/initial 24 hrs ACQUISITION ................................................ [ GPS 1PPS Valid ] Tracking: 6 ____ Not Tracking: 2 ________ Time _____ +1 leap second pending PRN El Az SS PRN El Az UTC 23:11:07 19 Dec 2016 2 31 69 83 6 15 34 GPS 1PPS Synchronized to UTC 12 41 80 110 32 27 234 ANT DLY 0 ns 14 37 266 81 Position ________________________ 25 83 33 127 MODE Hold 29 55 195 135 31 42 301 75 LAT N 51:39:04.198 LON E 0:46:36.358 ELEV MASK 10 deg HGT +45.42 m (MSL) HEALTH MONITOR ......................................................... [ OK ] Self Test: OK Int Pwr: OK Oven Pwr: OK OCXO: OK EFC: OK GPS Rcv: OK scpi >
CS
Charles Steinmetz
Tue, Dec 20, 2016 2:30 AM

David wrote:

This does appear now to be working - see below. I'm not exactly sure what
was the secret to getting this to work, but the above procedure seemed ok.

Time.  It is not at all unusual for a 58503A that has no almanac (or an
out-of-date almanac) to take several days to achieve disciplined
operation (and much longer for the OCXO to fully stabilize).  You may
get lucky and find that it happens faster, but don't count on it.  FWIW,
HP GPSDOs often take longer than others to initialize.

Everyone, please remember -- GPSDOs can take significant time to
re-initialize.  No need to post about what may be wrong until at least 3
days of continuous running after powering on.  Chill.

Same with respect to crystal oscillators settling to low drift rates.
No need to worry about drift (retrace) for at least a month (sometimes 6
or even 12 or 24 months) of continuous running after powering on.

Honestly, the best thing to do is turn it on and forget about it for 3
days (GPSDO) or 1 month (crystal oscillator).  DO NOT observe it (log
it, if you feel the need, and don't look at the log until after the 3/30
days).

Precise time and frequency require patience!

Best regards,

Charles

David wrote: > This does appear now to be working - see below. I'm not exactly sure what > was the secret to getting this to work, but the above procedure seemed ok. Time. It is not at all unusual for a 58503A that has no almanac (or an out-of-date almanac) to take several days to achieve disciplined operation (and much longer for the OCXO to fully stabilize). You may get lucky and find that it happens faster, but don't count on it. FWIW, HP GPSDOs often take longer than others to initialize. Everyone, please remember -- GPSDOs can take significant time to re-initialize. No need to post about what may be wrong until at least 3 days of continuous running after powering on. Chill. Same with respect to crystal oscillators settling to low drift rates. No need to worry about drift (retrace) for at least a month (sometimes 6 or even 12 or 24 months) of continuous running after powering on. Honestly, the best thing to do is turn it on and forget about it for 3 days (GPSDO) or 1 month (crystal oscillator). DO NOT observe it (log it, if you feel the need, and don't look at the log until after the 3/30 days). Precise time and frequency require patience! Best regards, Charles