how many seconds out does GPS discipline being to improve Rubidium stability?

GG
gkk gb
Sat, Feb 11, 2017 12:35 AM

Hello experts, I need a Rubidium frequency reference for my company, and wonder if I also need to GPS discipline it.

I characterize crystal-based OCXOs for ADEV, MTIE, and TDEV, and my longest measurement time is 100,000 seconds (28 hours).

I'm looking at this graph from SRS for PRS10,

http://www.thinksrs.com/assets/instr/PRS10/PRS10diag2LG.gif

and thinking that as long as I calibrate a Rubidium source annually, there's no need for a GPS (since it only appears to degrade stability). Is this true in general, or is the graph misleading me because it may be true here, but not always.

So my question, is a GPS necessary to discipline a Rubidium standard to characterize the best crystal oscillators for stability, or can I do without it (and just calibrate the Rubidium annually to maintain accuracy) and actually get better stability?

How many seconds out is a GPS generally needed to improve accuracy from a Rubidium standard?

Hello experts, I need a Rubidium frequency reference for my company, and wonder if I also need to GPS discipline it. I characterize crystal-based OCXOs for ADEV, MTIE, and TDEV, and my longest measurement time is 100,000 seconds (28 hours). I'm looking at this graph from SRS for PRS10, http://www.thinksrs.com/assets/instr/PRS10/PRS10diag2LG.gif and thinking that as long as I calibrate a Rubidium source annually, there's no need for a GPS (since it only appears to degrade stability). Is this true in general, or is the graph misleading me because it may be true here, but not always. So my question, is a GPS necessary to discipline a Rubidium standard to characterize the best crystal oscillators for stability, or can I do without it (and just calibrate the Rubidium annually to maintain accuracy) and actually get better stability? How many seconds out is a GPS generally needed to improve accuracy from a Rubidium standard?
DJ
David J Taylor
Sat, Feb 11, 2017 7:59 AM

From: gkk gb

Hello experts, I need a Rubidium frequency reference for my company, and
wonder if I also need to GPS discipline it.

I characterize crystal-based OCXOs for ADEV, MTIE, and TDEV, and my longest
measurement time is 100,000 seconds (28 hours).
[]

---===========

While you are waiting for your Rubidium, perhaps this little GPS box may
help?

http://www.leobodnar.com/shop/index.php?main_page=product_info&products_id=234

Cheers,
David

SatSignal Software - Quality software written to your requirements
Web: http://www.satsignal.eu
Email: david-taylor@blueyonder.co.uk
Twitter: @gm8arv

From: gkk gb Hello experts, I need a Rubidium frequency reference for my company, and wonder if I also need to GPS discipline it. I characterize crystal-based OCXOs for ADEV, MTIE, and TDEV, and my longest measurement time is 100,000 seconds (28 hours). [] ============================================ While you are waiting for your Rubidium, perhaps this little GPS box may help? http://www.leobodnar.com/shop/index.php?main_page=product_info&products_id=234 Cheers, David -- SatSignal Software - Quality software written to your requirements Web: http://www.satsignal.eu Email: david-taylor@blueyonder.co.uk Twitter: @gm8arv
BC
Bob Camp
Sat, Feb 11, 2017 2:29 PM

Hi

Backing up a bit here.

On Feb 10, 2017, at 7:35 PM, gkk gb modjklist@comcast.net wrote:

Hello experts, I need a Rubidium frequency reference for my company, and wonder if I also need to GPS discipline it.

I characterize crystal-based OCXOs for ADEV, MTIE, and TDEV, and my longest measurement time is 100,000 seconds (28 hours).

If your longest measurement is a 100,000 second ADEV, then your measurement time will be out in the
1,000,000 to 10,000,000 second range. Is that really what you are doing?

If 100,000 seconds ADEV is your longest measurement, what is the shortest tau you are interested in?
A Rb is not going to be much use for testing a good OCXO at shorter tau. Where the crossover happens
depends a lot on the grade of OCXO you are working with. By the time you get to 1 second
most OCXO’s will be noticeably better than most Rb’s.

I'm looking at this graph from SRS for PRS10,

http://www.thinksrs.com/assets/instr/PRS10/PRS10diag2LG.gif

I would suggest that plot is probably not the best one to depend on for GPS performance.  In a GPSDO setting
the cut over points are all over the place depending on which design you look at.

and thinking that as long as I calibrate a Rubidium source annually, there's no need for a GPS (since it only appears to degrade stability). Is this true in general, or is the graph misleading me because it may be true here, but not always.

The big issue is going to be temperature stability. If you have a Rb that is (say) 5x10^-10 over 0 to 50C, that is likely 1x10^-11 / C (or maybe more). A 2C delta in
your lab as the HVAC cycles will give you a 2x10^-11 “hump” in your ADEV plot.

Also consider that if you want an “easy” measurement of the devices you are testing, the reference source probably should be
5X better than what you expect out of the DUT. You probably will not have that luxury in this case. That gets you into multiple
references and things like three corner hat testing.

So my question, is a GPS necessary to discipline a Rubidium standard to characterize the best crystal oscillators for stability, or can I do without it (and just calibrate the Rubidium annually to maintain accuracy) and actually get better stability?

How many seconds out is a GPS generally needed to improve accuracy from a Rubidium standard?

If you really are running 1,000,000 to 10,000,000 second long tests, you need the GPS.

Lots of variables

Bob


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Backing up a bit here. > On Feb 10, 2017, at 7:35 PM, gkk gb <modjklist@comcast.net> wrote: > > Hello experts, I need a Rubidium frequency reference for my company, and wonder if I also need to GPS discipline it. > > > I characterize crystal-based OCXOs for ADEV, MTIE, and TDEV, and my longest measurement time is 100,000 seconds (28 hours). If your longest measurement is a 100,000 second ADEV, then your measurement time will be out in the 1,000,000 to 10,000,000 second range. Is that really what you are doing? If 100,000 seconds ADEV is your longest measurement, what is the shortest tau you are interested in? A Rb is not going to be much use for testing a good OCXO at shorter tau. Where the crossover happens depends a lot on the grade of OCXO you are working with. By the time you get to 1 second most OCXO’s will be noticeably better than most Rb’s. > > > I'm looking at this graph from SRS for PRS10, > > > http://www.thinksrs.com/assets/instr/PRS10/PRS10diag2LG.gif I would suggest that plot is probably not the best one to depend on for GPS performance. In a GPSDO setting the cut over points are all over the place depending on which design you look at. > > > and thinking that as long as I calibrate a Rubidium source annually, there's no need for a GPS (since it only appears to degrade stability). Is this true in general, or is the graph misleading me because it may be true here, but not always. The big issue is going to be temperature stability. If you have a Rb that is (say) 5x10^-10 over 0 to 50C, that is likely 1x10^-11 / C (or maybe more). A 2C delta in your lab as the HVAC cycles will give you a 2x10^-11 “hump” in your ADEV plot. Also consider that if you want an “easy” measurement of the devices you are testing, the reference source probably should be 5X better than what you expect out of the DUT. You probably will not have that luxury in this case. That gets you into multiple references and things like three corner hat testing. > > > So my question, is a GPS necessary to discipline a Rubidium standard to characterize the best crystal oscillators for stability, or can I do without it (and just calibrate the Rubidium annually to maintain accuracy) and actually get better stability? > > > How many seconds out is a GPS generally needed to improve accuracy from a Rubidium standard? If you really are running 1,000,000 to 10,000,000 second long tests, you need the GPS. Lots of variables Bob > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
GG
gkk gb
Sat, Feb 11, 2017 10:52 PM

Thanks Bob,

I should clarify the MTIE measurement extends 100000 seconds (the others are less time). Is it a reasonable question to ask if GPS is needed? Or are there other variables that are involved?

Good point about the temperature stability, I hadn't considered that. Can I place in a temperature chamber to provide a better thermal environment, or does that cause other issues (vibration from blowers, EMI noise, etc.)? Other ways to mitigate temperature changes?

It seems a Rubidium is good after a timescale of 100 s. What do people do below 100 s to characterize quartz oscillators. Do they simply try to find the most stable parts they can afford and break the x-axis (tau) into two regions using difference references for each? If so, are there generally accepted "gold" standards anyone can recommend for crystal products with the best stability to use as a reference between 0.1 and 100 seconds, for example?

 On February 11, 2017 at 6:29 AM Bob Camp <kb8tq@n1k.org> wrote:

 Hi

 Backing up a bit here.
     On Feb 10, 2017, at 7:35 PM, gkk gb <modjklist@comcast.net> wrote:

     Hello experts, I need a Rubidium frequency reference for my company, and wonder if I also need to GPS discipline it.

     I characterize crystal-based OCXOs for ADEV, MTIE, and TDEV, and my longest measurement time is 100,000 seconds (28 hours).
 If your longest measurement is a 100,000 second ADEV, then your measurement time will be out in the
 1,000,000 to 10,000,000 second range. Is that really what you are doing?

 If 100,000 seconds ADEV is your longest measurement, what is the shortest tau you are interested in?
 A Rb is not going to be much use for testing a good OCXO at shorter tau. Where the crossover happens
 depends a lot on the grade of OCXO you are working with. By the time you get to 1 second
 most OCXO’s will be noticeably better than most Rb’s.
     I'm looking at this graph from SRS for PRS10,

     http://www.thinksrs.com/assets/instr/PRS10/PRS10diag2LG.gif
 I would suggest that plot is probably not the best one to depend on for GPS performance. In a GPSDO setting
 the cut over points are all over the place depending on which design you look at.
     and thinking that as long as I calibrate a Rubidium source annually, there's no need for a GPS (since it only appears to degrade stability). Is this true in general, or is the graph misleading me because it may be true here, but not always.
 The big issue is going to be temperature stability. If you have a Rb that is (say) 5x10^-10 over 0 to 50C, that is likely 1x10^-11 / C (or maybe more). A 2C delta in
 your lab as the HVAC cycles will give you a 2x10^-11 “hump” in your ADEV plot.

 Also consider that if you want an “easy” measurement of the devices you are testing, the reference source probably should be
 5X better than what you expect out of the DUT. You probably will not have that luxury in this case. That gets you into multiple
 references and things like three corner hat testing.
     So my question, is a GPS necessary to discipline a Rubidium standard to characterize the best crystal oscillators for stability, or can I do without it (and just calibrate the Rubidium annually to maintain accuracy) and actually get better stability?

     How many seconds out is a GPS generally needed to improve accuracy from a Rubidium standard?
 If you really are running 1,000,000 to 10,000,000 second long tests, you need the GPS.

 Lots of variables

 Bob
     _______________________________________________
     time-nuts mailing list -- time-nuts@febo.com
     To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
     and follow the instructions there.
Thanks Bob, I should clarify the MTIE measurement extends 100000 seconds (the others are less time). Is it a reasonable question to ask if GPS is needed? Or are there other variables that are involved? Good point about the temperature stability, I hadn't considered that. Can I place in a temperature chamber to provide a better thermal environment, or does that cause other issues (vibration from blowers, EMI noise, etc.)? Other ways to mitigate temperature changes? It seems a Rubidium is good after a timescale of 100 s. What do people do below 100 s to characterize quartz oscillators. Do they simply try to find the most stable parts they can afford and break the x-axis (tau) into two regions using difference references for each? If so, are there generally accepted "gold" standards anyone can recommend for crystal products with the best stability to use as a reference between 0.1 and 100 seconds, for example? > > On February 11, 2017 at 6:29 AM Bob Camp <kb8tq@n1k.org> wrote: > > Hi > > Backing up a bit here. > > > > > > On Feb 10, 2017, at 7:35 PM, gkk gb <modjklist@comcast.net> wrote: > > > > Hello experts, I need a Rubidium frequency reference for my company, and wonder if I also need to GPS discipline it. > > > > I characterize crystal-based OCXOs for ADEV, MTIE, and TDEV, and my longest measurement time is 100,000 seconds (28 hours). > > > > > > If your longest measurement is a 100,000 second ADEV, then your measurement time will be out in the > 1,000,000 to 10,000,000 second range. Is that really what you are doing? > > If 100,000 seconds ADEV is your longest measurement, what is the shortest tau you are interested in? > A Rb is not going to be much use for testing a good OCXO at shorter tau. Where the crossover happens > depends a lot on the grade of OCXO you are working with. By the time you get to 1 second > most OCXO’s will be noticeably better than most Rb’s. > > > > > > I'm looking at this graph from SRS for PRS10, > > > > http://www.thinksrs.com/assets/instr/PRS10/PRS10diag2LG.gif > > > > > > I would suggest that plot is probably not the best one to depend on for GPS performance. In a GPSDO setting > the cut over points are all over the place depending on which design you look at. > > > > > > and thinking that as long as I calibrate a Rubidium source annually, there's no need for a GPS (since it only appears to degrade stability). Is this true in general, or is the graph misleading me because it may be true here, but not always. > > > > > > The big issue is going to be temperature stability. If you have a Rb that is (say) 5x10^-10 over 0 to 50C, that is likely 1x10^-11 / C (or maybe more). A 2C delta in > your lab as the HVAC cycles will give you a 2x10^-11 “hump” in your ADEV plot. > > Also consider that if you want an “easy” measurement of the devices you are testing, the reference source probably should be > 5X better than what you expect out of the DUT. You probably will not have that luxury in this case. That gets you into multiple > references and things like three corner hat testing. > > > > > > So my question, is a GPS necessary to discipline a Rubidium standard to characterize the best crystal oscillators for stability, or can I do without it (and just calibrate the Rubidium annually to maintain accuracy) and actually get better stability? > > > > How many seconds out is a GPS generally needed to improve accuracy from a Rubidium standard? > > > > > > If you really are running 1,000,000 to 10,000,000 second long tests, you need the GPS. > > Lots of variables > > Bob > > > > > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > > > > >
BC
Bob Camp
Sat, Feb 11, 2017 11:52 PM

Hi

Using ADEV as an example (the other stuff will have it’s own curves, but the result is the same):

A typical Rb should have a stability at short tau that goes as 1/ square root(Tau). If you are at 2x10^-11 at 1 second, you
will be at 2x10^-12 at 100 seconds and 2x10^-13 at 1,000 seconds. Somewhere in the parts in 10^-13 that relation will
start to diverge from reality.

A fairly normal low frequency OCXO has a stability that is fairly flat with tau in the 1 to 100 second range. If they have been
on power constantly that “flat zone" may extend to 1,000 seconds. Floors should be in the low parts in 10^-12 to mid parts
in 10^-13 range.

A good OCXO may beat a normal Rb at 1,000 seconds. That may or may not be an issue in your case. It depends a lot
on what you are trying to do.

Simple solutions:

  1. Run something better than an Rb. A hydrogen maser is one alternative (simple if you don’t have to pay for it).
  2. Do all your measurements as three corner hats. You run two references and one DUT into gear that will do that sort of test.
  3. Segment the measurements and use carefully selected references for those ranges.

None of those are actually simple. Number 3 sounds cool until you realize that you are switching test setups around a lot and
the devices you are using still need a setup like 2 to figure out which ones to use.

So do you need a GPS? What are the limits on your MTIE tests? (MTIE on an OCXO is highly dependent on several
things so there is no simple number there). A very normal quartz based GPSDO might be a fine reference for your test.

How much shorter are the other tests? Is ADEV at 1,000 seconds even of interest? If the answer is < 1,000 seconds a
Rb may not do you much good at all.

Lots of twists and turns.

Bob

On Feb 11, 2017, at 5:52 PM, gkk gb modjklist@comcast.net wrote:

Thanks Bob,

I should clarify the MTIE measurement extends 100000 seconds (the others are less time). Is it a reasonable question to ask if GPS is needed? Or are there other variables that are involved?

Good point about the temperature stability, I hadn't considered that. Can I place in a temperature chamber to provide a better thermal environment, or does that cause other issues (vibration from blowers, EMI noise, etc.)? Other ways to mitigate temperature changes?

It seems a Rubidium is good after a timescale of 100 s. What do people do below 100 s to characterize quartz oscillators. Do they simply try to find the most stable parts they can afford and break the x-axis (tau) into two regions using difference references for each? If so, are there generally accepted "gold" standards anyone can recommend for crystal products with the best stability to use as a reference between 0.1 and 100 seconds, for example?

On February 11, 2017 at 6:29 AM Bob Camp kb8tq@n1k.org wrote:

Hi

Backing up a bit here.

On Feb 10, 2017, at 7:35 PM, gkk gb modjklist@comcast.net wrote:

Hello experts, I need a Rubidium frequency reference for my company, and wonder if I also need to GPS discipline it.

I characterize crystal-based OCXOs for ADEV, MTIE, and TDEV, and my longest measurement time is 100,000 seconds (28 hours).

If your longest measurement is a 100,000 second ADEV, then your measurement time will be out in the
1,000,000 to 10,000,000 second range. Is that really what you are doing?

If 100,000 seconds ADEV is your longest measurement, what is the shortest tau you are interested in?
A Rb is not going to be much use for testing a good OCXO at shorter tau. Where the crossover happens
depends a lot on the grade of OCXO you are working with. By the time you get to 1 second
most OCXO’s will be noticeably better than most Rb’s.

I'm looking at this graph from SRS for PRS10,

http://www.thinksrs.com/assets/instr/PRS10/PRS10diag2LG.gif

I would suggest that plot is probably not the best one to depend on for GPS performance. In a GPSDO setting
the cut over points are all over the place depending on which design you look at.

and thinking that as long as I calibrate a Rubidium source annually, there's no need for a GPS (since it only appears to degrade stability). Is this true in general, or is the graph misleading me because it may be true here, but not always.

The big issue is going to be temperature stability. If you have a Rb that is (say) 5x10^-10 over 0 to 50C, that is likely 1x10^-11 / C (or maybe more). A 2C delta in
your lab as the HVAC cycles will give you a 2x10^-11 “hump” in your ADEV plot.

Also consider that if you want an “easy” measurement of the devices you are testing, the reference source probably should be
5X better than what you expect out of the DUT. You probably will not have that luxury in this case. That gets you into multiple
references and things like three corner hat testing.

So my question, is a GPS necessary to discipline a Rubidium standard to characterize the best crystal oscillators for stability, or can I do without it (and just calibrate the Rubidium annually to maintain accuracy) and actually get better stability?

How many seconds out is a GPS generally needed to improve accuracy from a Rubidium standard?

If you really are running 1,000,000 to 10,000,000 second long tests, you need the GPS.

Lots of variables

Bob


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Using ADEV as an example (the other stuff will have it’s own curves, but the result is the same): A typical Rb should have a stability at short tau that goes as 1/ square root(Tau). If you are at 2x10^-11 at 1 second, you will be at 2x10^-12 at 100 seconds and 2x10^-13 at 1,000 seconds. Somewhere in the parts in 10^-13 that relation will start to diverge from reality. A fairly normal low frequency OCXO has a stability that is fairly flat with tau in the 1 to 100 second range. If they have been on power constantly that “flat zone" may extend to 1,000 seconds. Floors should be in the low parts in 10^-12 to mid parts in 10^-13 range. A good OCXO *may* beat a normal Rb at 1,000 seconds. That may or may not be an issue in your case. It depends a lot on what you are trying to do. Simple solutions: 1) Run something better than an Rb. A hydrogen maser is one alternative (simple if you don’t have to pay for it). 2) Do all your measurements as three corner hats. You run two references and one DUT into gear that will do that sort of test. 3) Segment the measurements and use carefully selected references for those ranges. None of those are actually simple. Number 3 sounds cool until you realize that you are switching test setups around a lot and the devices you are using still need a setup like 2 to figure out which ones to use. So do you need a GPS? What are the limits on your MTIE tests? (MTIE on an OCXO is highly dependent on several things so there is no simple number there). A very normal quartz based GPSDO might be a fine reference for your test. How much shorter are the other tests? Is ADEV at 1,000 seconds even of interest? If the answer is < 1,000 seconds a Rb may not do you much good at all. Lots of twists and turns. Bob > On Feb 11, 2017, at 5:52 PM, gkk gb <modjklist@comcast.net> wrote: > > Thanks Bob, > > > > I should clarify the MTIE measurement extends 100000 seconds (the others are less time). Is it a reasonable question to ask if GPS is needed? Or are there other variables that are involved? > > > > Good point about the temperature stability, I hadn't considered that. Can I place in a temperature chamber to provide a better thermal environment, or does that cause other issues (vibration from blowers, EMI noise, etc.)? Other ways to mitigate temperature changes? > > > > It seems a Rubidium is good after a timescale of 100 s. What do people do below 100 s to characterize quartz oscillators. Do they simply try to find the most stable parts they can afford and break the x-axis (tau) into two regions using difference references for each? If so, are there generally accepted "gold" standards anyone can recommend for crystal products with the best stability to use as a reference between 0.1 and 100 seconds, for example? > > > > On February 11, 2017 at 6:29 AM Bob Camp <kb8tq@n1k.org> wrote: > > Hi > > Backing up a bit here. > > On Feb 10, 2017, at 7:35 PM, gkk gb <modjklist@comcast.net> wrote: > > Hello experts, I need a Rubidium frequency reference for my company, and wonder if I also need to GPS discipline it. > > I characterize crystal-based OCXOs for ADEV, MTIE, and TDEV, and my longest measurement time is 100,000 seconds (28 hours). > > If your longest measurement is a 100,000 second ADEV, then your measurement time will be out in the > 1,000,000 to 10,000,000 second range. Is that really what you are doing? > > If 100,000 seconds ADEV is your longest measurement, what is the shortest tau you are interested in? > A Rb is not going to be much use for testing a good OCXO at shorter tau. Where the crossover happens > depends a lot on the grade of OCXO you are working with. By the time you get to 1 second > most OCXO’s will be noticeably better than most Rb’s. > > I'm looking at this graph from SRS for PRS10, > > http://www.thinksrs.com/assets/instr/PRS10/PRS10diag2LG.gif > > I would suggest that plot is probably not the best one to depend on for GPS performance. In a GPSDO setting > the cut over points are all over the place depending on which design you look at. > > and thinking that as long as I calibrate a Rubidium source annually, there's no need for a GPS (since it only appears to degrade stability). Is this true in general, or is the graph misleading me because it may be true here, but not always. > > The big issue is going to be temperature stability. If you have a Rb that is (say) 5x10^-10 over 0 to 50C, that is likely 1x10^-11 / C (or maybe more). A 2C delta in > your lab as the HVAC cycles will give you a 2x10^-11 “hump” in your ADEV plot. > > Also consider that if you want an “easy” measurement of the devices you are testing, the reference source probably should be > 5X better than what you expect out of the DUT. You probably will not have that luxury in this case. That gets you into multiple > references and things like three corner hat testing. > > So my question, is a GPS necessary to discipline a Rubidium standard to characterize the best crystal oscillators for stability, or can I do without it (and just calibrate the Rubidium annually to maintain accuracy) and actually get better stability? > > How many seconds out is a GPS generally needed to improve accuracy from a Rubidium standard? > > If you really are running 1,000,000 to 10,000,000 second long tests, you need the GPS. > > Lots of variables > > Bob > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
PR
Peter Reilley
Sun, Feb 12, 2017 12:12 PM

Could you do all three tests in parallel?  One unit under test driving
three counters.
Each counter using a different reference signal, one on a OCXO, one on a
rubidium,
and one on a GPS disciplined oscillator.  At each point in time during
the test simply
choose the one that gives the best ADEV?

Pete.

On 2/11/2017 6:52 PM, Bob Camp wrote:

Hi

Using ADEV as an example (the other stuff will have it’s own curves, but the result is the same):

A typical Rb should have a stability at short tau that goes as 1/ square root(Tau). If you are at 2x10^-11 at 1 second, you
will be at 2x10^-12 at 100 seconds and 2x10^-13 at 1,000 seconds. Somewhere in the parts in 10^-13 that relation will
start to diverge from reality.

A fairly normal low frequency OCXO has a stability that is fairly flat with tau in the 1 to 100 second range. If they have been
on power constantly that “flat zone" may extend to 1,000 seconds. Floors should be in the low parts in 10^-12 to mid parts
in 10^-13 range.

A good OCXO may beat a normal Rb at 1,000 seconds. That may or may not be an issue in your case. It depends a lot
on what you are trying to do.

Simple solutions:

  1. Run something better than an Rb. A hydrogen maser is one alternative (simple if you don’t have to pay for it).
  2. Do all your measurements as three corner hats. You run two references and one DUT into gear that will do that sort of test.
  3. Segment the measurements and use carefully selected references for those ranges.

None of those are actually simple. Number 3 sounds cool until you realize that you are switching test setups around a lot and
the devices you are using still need a setup like 2 to figure out which ones to use.

So do you need a GPS? What are the limits on your MTIE tests? (MTIE on an OCXO is highly dependent on several
things so there is no simple number there). A very normal quartz based GPSDO might be a fine reference for your test.

How much shorter are the other tests? Is ADEV at 1,000 seconds even of interest? If the answer is < 1,000 seconds a
Rb may not do you much good at all.

Lots of twists and turns.

Bob

On Feb 11, 2017, at 5:52 PM, gkk gb modjklist@comcast.net wrote:

Thanks Bob,

I should clarify the MTIE measurement extends 100000 seconds (the others are less time). Is it a reasonable question to ask if GPS is needed? Or are there other variables that are involved?

Good point about the temperature stability, I hadn't considered that. Can I place in a temperature chamber to provide a better thermal environment, or does that cause other issues (vibration from blowers, EMI noise, etc.)? Other ways to mitigate temperature changes?

It seems a Rubidium is good after a timescale of 100 s. What do people do below 100 s to characterize quartz oscillators. Do they simply try to find the most stable parts they can afford and break the x-axis (tau) into two regions using difference references for each? If so, are there generally accepted "gold" standards anyone can recommend for crystal products with the best stability to use as a reference between 0.1 and 100 seconds, for example?

On February 11, 2017 at 6:29 AM Bob Camp kb8tq@n1k.org wrote:

Hi

Backing up a bit here.

On Feb 10, 2017, at 7:35 PM, gkk gb modjklist@comcast.net wrote:

Hello experts, I need a Rubidium frequency reference for my company, and wonder if I also need to GPS discipline it.

I characterize crystal-based OCXOs for ADEV, MTIE, and TDEV, and my longest measurement time is 100,000 seconds (28 hours).

If your longest measurement is a 100,000 second ADEV, then your measurement time will be out in the
1,000,000 to 10,000,000 second range. Is that really what you are doing?

If 100,000 seconds ADEV is your longest measurement, what is the shortest tau you are interested in?
A Rb is not going to be much use for testing a good OCXO at shorter tau. Where the crossover happens
depends a lot on the grade of OCXO you are working with. By the time you get to 1 second
most OCXO’s will be noticeably better than most Rb’s.

I'm looking at this graph from SRS for PRS10,

http://www.thinksrs.com/assets/instr/PRS10/PRS10diag2LG.gif

I would suggest that plot is probably not the best one to depend on for GPS performance. In a GPSDO setting
the cut over points are all over the place depending on which design you look at.

and thinking that as long as I calibrate a Rubidium source annually, there's no need for a GPS (since it only appears to degrade stability). Is this true in general, or is the graph misleading me because it may be true here, but not always.

The big issue is going to be temperature stability. If you have a Rb that is (say) 5x10^-10 over 0 to 50C, that is likely 1x10^-11 / C (or maybe more). A 2C delta in
your lab as the HVAC cycles will give you a 2x10^-11 “hump” in your ADEV plot.

Also consider that if you want an “easy” measurement of the devices you are testing, the reference source probably should be
5X better than what you expect out of the DUT. You probably will not have that luxury in this case. That gets you into multiple
references and things like three corner hat testing.

So my question, is a GPS necessary to discipline a Rubidium standard to characterize the best crystal oscillators for stability, or can I do without it (and just calibrate the Rubidium annually to maintain accuracy) and actually get better stability?

How many seconds out is a GPS generally needed to improve accuracy from a Rubidium standard?

If you really are running 1,000,000 to 10,000,000 second long tests, you need the GPS.

Lots of variables

Bob


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Could you do all three tests in parallel? One unit under test driving three counters. Each counter using a different reference signal, one on a OCXO, one on a rubidium, and one on a GPS disciplined oscillator. At each point in time during the test simply choose the one that gives the best ADEV? Pete. On 2/11/2017 6:52 PM, Bob Camp wrote: > Hi > > Using ADEV as an example (the other stuff will have it’s own curves, but the result is the same): > > A typical Rb should have a stability at short tau that goes as 1/ square root(Tau). If you are at 2x10^-11 at 1 second, you > will be at 2x10^-12 at 100 seconds and 2x10^-13 at 1,000 seconds. Somewhere in the parts in 10^-13 that relation will > start to diverge from reality. > > A fairly normal low frequency OCXO has a stability that is fairly flat with tau in the 1 to 100 second range. If they have been > on power constantly that “flat zone" may extend to 1,000 seconds. Floors should be in the low parts in 10^-12 to mid parts > in 10^-13 range. > > A good OCXO *may* beat a normal Rb at 1,000 seconds. That may or may not be an issue in your case. It depends a lot > on what you are trying to do. > > Simple solutions: > > 1) Run something better than an Rb. A hydrogen maser is one alternative (simple if you don’t have to pay for it). > 2) Do all your measurements as three corner hats. You run two references and one DUT into gear that will do that sort of test. > 3) Segment the measurements and use carefully selected references for those ranges. > > None of those are actually simple. Number 3 sounds cool until you realize that you are switching test setups around a lot and > the devices you are using still need a setup like 2 to figure out which ones to use. > > So do you need a GPS? What are the limits on your MTIE tests? (MTIE on an OCXO is highly dependent on several > things so there is no simple number there). A very normal quartz based GPSDO might be a fine reference for your test. > > How much shorter are the other tests? Is ADEV at 1,000 seconds even of interest? If the answer is < 1,000 seconds a > Rb may not do you much good at all. > > Lots of twists and turns. > > Bob > > > > > >> On Feb 11, 2017, at 5:52 PM, gkk gb <modjklist@comcast.net> wrote: >> >> Thanks Bob, >> >> >> >> I should clarify the MTIE measurement extends 100000 seconds (the others are less time). Is it a reasonable question to ask if GPS is needed? Or are there other variables that are involved? >> >> >> >> Good point about the temperature stability, I hadn't considered that. Can I place in a temperature chamber to provide a better thermal environment, or does that cause other issues (vibration from blowers, EMI noise, etc.)? Other ways to mitigate temperature changes? >> >> >> >> It seems a Rubidium is good after a timescale of 100 s. What do people do below 100 s to characterize quartz oscillators. Do they simply try to find the most stable parts they can afford and break the x-axis (tau) into two regions using difference references for each? If so, are there generally accepted "gold" standards anyone can recommend for crystal products with the best stability to use as a reference between 0.1 and 100 seconds, for example? >> >> >> >> On February 11, 2017 at 6:29 AM Bob Camp <kb8tq@n1k.org> wrote: >> >> Hi >> >> Backing up a bit here. >> >> On Feb 10, 2017, at 7:35 PM, gkk gb <modjklist@comcast.net> wrote: >> >> Hello experts, I need a Rubidium frequency reference for my company, and wonder if I also need to GPS discipline it. >> >> I characterize crystal-based OCXOs for ADEV, MTIE, and TDEV, and my longest measurement time is 100,000 seconds (28 hours). >> >> If your longest measurement is a 100,000 second ADEV, then your measurement time will be out in the >> 1,000,000 to 10,000,000 second range. Is that really what you are doing? >> >> If 100,000 seconds ADEV is your longest measurement, what is the shortest tau you are interested in? >> A Rb is not going to be much use for testing a good OCXO at shorter tau. Where the crossover happens >> depends a lot on the grade of OCXO you are working with. By the time you get to 1 second >> most OCXO’s will be noticeably better than most Rb’s. >> >> I'm looking at this graph from SRS for PRS10, >> >> http://www.thinksrs.com/assets/instr/PRS10/PRS10diag2LG.gif >> >> I would suggest that plot is probably not the best one to depend on for GPS performance. In a GPSDO setting >> the cut over points are all over the place depending on which design you look at. >> >> and thinking that as long as I calibrate a Rubidium source annually, there's no need for a GPS (since it only appears to degrade stability). Is this true in general, or is the graph misleading me because it may be true here, but not always. >> >> The big issue is going to be temperature stability. If you have a Rb that is (say) 5x10^-10 over 0 to 50C, that is likely 1x10^-11 / C (or maybe more). A 2C delta in >> your lab as the HVAC cycles will give you a 2x10^-11 “hump” in your ADEV plot. >> >> Also consider that if you want an “easy” measurement of the devices you are testing, the reference source probably should be >> 5X better than what you expect out of the DUT. You probably will not have that luxury in this case. That gets you into multiple >> references and things like three corner hat testing. >> >> So my question, is a GPS necessary to discipline a Rubidium standard to characterize the best crystal oscillators for stability, or can I do without it (and just calibrate the Rubidium annually to maintain accuracy) and actually get better stability? >> >> How many seconds out is a GPS generally needed to improve accuracy from a Rubidium standard? >> >> If you really are running 1,000,000 to 10,000,000 second long tests, you need the GPS. >> >> Lots of variables >> >> Bob >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
GG
gkk gb
Sun, Feb 12, 2017 6:16 PM

That is an interesting suggestion (thanks), and would indeed work for me if it is possible to split the DUT signal into 3 signals in a such a way that wouldn't affect measurements for ADEV, TDEV, MTIE. But I'm thinking anything active would introduce it's own noise into the signal and change the data. Has this technique been used to success in the past?

On February 12, 2017 at 4:12 AM Peter Reilley preilley_454@comcast.net wrote:

Could you do all three tests in parallel?  One unit under test driving
three counters.
Each counter using a different reference signal, one on a OCXO, one on a
rubidium,
and one on a GPS disciplined oscillator.  At each point in time during
the test simply
choose the one that gives the best ADEV?

Pete.

On 2/11/2017 6:52 PM, Bob Camp wrote:

Hi

Using ADEV as an example (the other stuff will have it’s own curves, but the result is the same):

A typical Rb should have a stability at short tau that goes as 1/ square root(Tau). If you are at 2x10^-11 at 1 second, you
will be at 2x10^-12 at 100 seconds and 2x10^-13 at 1,000 seconds. Somewhere in the parts in 10^-13 that relation will
start to diverge from reality.

A fairly normal low frequency OCXO has a stability that is fairly flat with tau in the 1 to 100 second range. If they have been
on power constantly that “flat zone" may extend to 1,000 seconds. Floors should be in the low parts in 10^-12 to mid parts
in 10^-13 range.

A good OCXO may beat a normal Rb at 1,000 seconds. That may or may not be an issue in your case. It depends a lot
on what you are trying to do.

Simple solutions:

  1. Run something better than an Rb. A hydrogen maser is one alternative (simple if you don’t have to pay for it).
  2. Do all your measurements as three corner hats. You run two references and one DUT into gear that will do that sort of test.
  3. Segment the measurements and use carefully selected references for those ranges.

None of those are actually simple. Number 3 sounds cool until you realize that you are switching test setups around a lot and
the devices you are using still need a setup like 2 to figure out which ones to use.

So do you need a GPS? What are the limits on your MTIE tests? (MTIE on an OCXO is highly dependent on several
things so there is no simple number there). A very normal quartz based GPSDO might be a fine reference for your test.

How much shorter are the other tests? Is ADEV at 1,000 seconds even of interest? If the answer is < 1,000 seconds a
Rb may not do you much good at all.

Lots of twists and turns.

Bob

On Feb 11, 2017, at 5:52 PM, gkk gb modjklist@comcast.net wrote:

Thanks Bob,

I should clarify the MTIE measurement extends 100000 seconds (the others are less time). Is it a reasonable question to ask if GPS is needed? Or are there other variables that are involved?

Good point about the temperature stability, I hadn't considered that. Can I place in a temperature chamber to provide a better thermal environment, or does that cause other issues (vibration from blowers, EMI noise, etc.)? Other ways to mitigate temperature changes?

It seems a Rubidium is good after a timescale of 100 s. What do people do below 100 s to characterize quartz oscillators. Do they simply try to find the most stable parts they can afford and break the x-axis (tau) into two regions using difference references for each? If so, are there generally accepted "gold" standards anyone can recommend for crystal products with the best stability to use as a reference between 0.1 and 100 seconds, for example?

On February 11, 2017 at 6:29 AM Bob Camp kb8tq@n1k.org wrote:

Hi

Backing up a bit here.

On Feb 10, 2017, at 7:35 PM, gkk gb modjklist@comcast.net wrote:

Hello experts, I need a Rubidium frequency reference for my company, and wonder if I also need to GPS discipline it.

I characterize crystal-based OCXOs for ADEV, MTIE, and TDEV, and my longest measurement time is 100,000 seconds (28 hours).

If your longest measurement is a 100,000 second ADEV, then your measurement time will be out in the
1,000,000 to 10,000,000 second range. Is that really what you are doing?

If 100,000 seconds ADEV is your longest measurement, what is the shortest tau you are interested in?
A Rb is not going to be much use for testing a good OCXO at shorter tau. Where the crossover happens
depends a lot on the grade of OCXO you are working with. By the time you get to 1 second
most OCXO’s will be noticeably better than most Rb’s.

I'm looking at this graph from SRS for PRS10,

http://www.thinksrs.com/assets/instr/PRS10/PRS10diag2LG.gif

I would suggest that plot is probably not the best one to depend on for GPS performance. In a GPSDO setting
the cut over points are all over the place depending on which design you look at.

and thinking that as long as I calibrate a Rubidium source annually, there's no need for a GPS (since it only appears to degrade stability). Is this true in general, or is the graph misleading me because it may be true here, but not always.

The big issue is going to be temperature stability. If you have a Rb that is (say) 5x10^-10 over 0 to 50C, that is likely 1x10^-11 / C (or maybe more). A 2C delta in
your lab as the HVAC cycles will give you a 2x10^-11 “hump” in your ADEV plot.

Also consider that if you want an “easy” measurement of the devices you are testing, the reference source probably should be
5X better than what you expect out of the DUT. You probably will not have that luxury in this case. That gets you into multiple
references and things like three corner hat testing.

So my question, is a GPS necessary to discipline a Rubidium standard to characterize the best crystal oscillators for stability, or can I do without it (and just calibrate the Rubidium annually to maintain accuracy) and actually get better stability?

How many seconds out is a GPS generally needed to improve accuracy from a Rubidium standard?

If you really are running 1,000,000 to 10,000,000 second long tests, you need the GPS.

Lots of variables

Bob


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

That is an interesting suggestion (thanks), and would indeed work for me if it is possible to split the DUT signal into 3 signals in a such a way that wouldn't affect measurements for ADEV, TDEV, MTIE. But I'm thinking anything active would introduce it's own noise into the signal and change the data. Has this technique been used to success in the past? > On February 12, 2017 at 4:12 AM Peter Reilley <preilley_454@comcast.net> wrote: > > > Could you do all three tests in parallel? One unit under test driving > three counters. > Each counter using a different reference signal, one on a OCXO, one on a > rubidium, > and one on a GPS disciplined oscillator. At each point in time during > the test simply > choose the one that gives the best ADEV? > > Pete. > > On 2/11/2017 6:52 PM, Bob Camp wrote: > > Hi > > > > Using ADEV as an example (the other stuff will have it’s own curves, but the result is the same): > > > > A typical Rb should have a stability at short tau that goes as 1/ square root(Tau). If you are at 2x10^-11 at 1 second, you > > will be at 2x10^-12 at 100 seconds and 2x10^-13 at 1,000 seconds. Somewhere in the parts in 10^-13 that relation will > > start to diverge from reality. > > > > A fairly normal low frequency OCXO has a stability that is fairly flat with tau in the 1 to 100 second range. If they have been > > on power constantly that “flat zone" may extend to 1,000 seconds. Floors should be in the low parts in 10^-12 to mid parts > > in 10^-13 range. > > > > A good OCXO *may* beat a normal Rb at 1,000 seconds. That may or may not be an issue in your case. It depends a lot > > on what you are trying to do. > > > > Simple solutions: > > > > 1) Run something better than an Rb. A hydrogen maser is one alternative (simple if you don’t have to pay for it). > > 2) Do all your measurements as three corner hats. You run two references and one DUT into gear that will do that sort of test. > > 3) Segment the measurements and use carefully selected references for those ranges. > > > > None of those are actually simple. Number 3 sounds cool until you realize that you are switching test setups around a lot and > > the devices you are using still need a setup like 2 to figure out which ones to use. > > > > So do you need a GPS? What are the limits on your MTIE tests? (MTIE on an OCXO is highly dependent on several > > things so there is no simple number there). A very normal quartz based GPSDO might be a fine reference for your test. > > > > How much shorter are the other tests? Is ADEV at 1,000 seconds even of interest? If the answer is < 1,000 seconds a > > Rb may not do you much good at all. > > > > Lots of twists and turns. > > > > Bob > > > > > > > > > > > >> On Feb 11, 2017, at 5:52 PM, gkk gb <modjklist@comcast.net> wrote: > >> > >> Thanks Bob, > >> > >> > >> > >> I should clarify the MTIE measurement extends 100000 seconds (the others are less time). Is it a reasonable question to ask if GPS is needed? Or are there other variables that are involved? > >> > >> > >> > >> Good point about the temperature stability, I hadn't considered that. Can I place in a temperature chamber to provide a better thermal environment, or does that cause other issues (vibration from blowers, EMI noise, etc.)? Other ways to mitigate temperature changes? > >> > >> > >> > >> It seems a Rubidium is good after a timescale of 100 s. What do people do below 100 s to characterize quartz oscillators. Do they simply try to find the most stable parts they can afford and break the x-axis (tau) into two regions using difference references for each? If so, are there generally accepted "gold" standards anyone can recommend for crystal products with the best stability to use as a reference between 0.1 and 100 seconds, for example? > >> > >> > >> > >> On February 11, 2017 at 6:29 AM Bob Camp <kb8tq@n1k.org> wrote: > >> > >> Hi > >> > >> Backing up a bit here. > >> > >> On Feb 10, 2017, at 7:35 PM, gkk gb <modjklist@comcast.net> wrote: > >> > >> Hello experts, I need a Rubidium frequency reference for my company, and wonder if I also need to GPS discipline it. > >> > >> I characterize crystal-based OCXOs for ADEV, MTIE, and TDEV, and my longest measurement time is 100,000 seconds (28 hours). > >> > >> If your longest measurement is a 100,000 second ADEV, then your measurement time will be out in the > >> 1,000,000 to 10,000,000 second range. Is that really what you are doing? > >> > >> If 100,000 seconds ADEV is your longest measurement, what is the shortest tau you are interested in? > >> A Rb is not going to be much use for testing a good OCXO at shorter tau. Where the crossover happens > >> depends a lot on the grade of OCXO you are working with. By the time you get to 1 second > >> most OCXO’s will be noticeably better than most Rb’s. > >> > >> I'm looking at this graph from SRS for PRS10, > >> > >> http://www.thinksrs.com/assets/instr/PRS10/PRS10diag2LG.gif > >> > >> I would suggest that plot is probably not the best one to depend on for GPS performance. In a GPSDO setting > >> the cut over points are all over the place depending on which design you look at. > >> > >> and thinking that as long as I calibrate a Rubidium source annually, there's no need for a GPS (since it only appears to degrade stability). Is this true in general, or is the graph misleading me because it may be true here, but not always. > >> > >> The big issue is going to be temperature stability. If you have a Rb that is (say) 5x10^-10 over 0 to 50C, that is likely 1x10^-11 / C (or maybe more). A 2C delta in > >> your lab as the HVAC cycles will give you a 2x10^-11 “hump” in your ADEV plot. > >> > >> Also consider that if you want an “easy” measurement of the devices you are testing, the reference source probably should be > >> 5X better than what you expect out of the DUT. You probably will not have that luxury in this case. That gets you into multiple > >> references and things like three corner hat testing. > >> > >> So my question, is a GPS necessary to discipline a Rubidium standard to characterize the best crystal oscillators for stability, or can I do without it (and just calibrate the Rubidium annually to maintain accuracy) and actually get better stability? > >> > >> How many seconds out is a GPS generally needed to improve accuracy from a Rubidium standard? > >> > >> If you really are running 1,000,000 to 10,000,000 second long tests, you need the GPS. > >> > >> Lots of variables > >> > >> Bob > >> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@febo.com > >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > >> and follow the instructions there. > >> > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BC
Bob Camp
Sun, Feb 12, 2017 6:53 PM

Hi

You likely need to do this with something like a TimePod to get the resolution. Setting up
three of them (or anything similar) is going to cost you way more than the Rb.

If you have a 100 ps counter, you will get 1x10^-10 at 1 second for ADEV data. That will
go to 1x10^-11 at 10 seconds and 1x10^-12 at 100 seconds. You only hit 1x10^-13 at
1,000 seconds. Again it’s a “resolution 5X better than measurement” sort of thing. You
don’t want to read the floor of your test gear, you want to see the device under test. If
you have a 20 ps counter the numbers above would be your 5X limits rather than the floor.

Again, this gets back to the “what do you want to do?” and “how short is your shortest tau?”
questions. In the actual use case, there may be no interest at all in Tau’s shorter than
1,000 seconds. It’s also possible that the OCXO’s have an expected floor at 1x10^-10 from
1 second to 1,000 seconds. There are lots of variables.

Bob

On Feb 12, 2017, at 7:12 AM, Peter Reilley preilley_454@comcast.net wrote:

Could you do all three tests in parallel?  One unit under test driving three counters.
Each counter using a different reference signal, one on a OCXO, one on a rubidium,
and one on a GPS disciplined oscillator.  At each point in time during the test simply
choose the one that gives the best ADEV?

Pete.

On 2/11/2017 6:52 PM, Bob Camp wrote:

Hi

Using ADEV as an example (the other stuff will have it’s own curves, but the result is the same):

A typical Rb should have a stability at short tau that goes as 1/ square root(Tau). If you are at 2x10^-11 at 1 second, you
will be at 2x10^-12 at 100 seconds and 2x10^-13 at 1,000 seconds. Somewhere in the parts in 10^-13 that relation will
start to diverge from reality.

A fairly normal low frequency OCXO has a stability that is fairly flat with tau in the 1 to 100 second range. If they have been
on power constantly that “flat zone" may extend to 1,000 seconds. Floors should be in the low parts in 10^-12 to mid parts
in 10^-13 range.

A good OCXO may beat a normal Rb at 1,000 seconds. That may or may not be an issue in your case. It depends a lot
on what you are trying to do.

Simple solutions:

  1. Run something better than an Rb. A hydrogen maser is one alternative (simple if you don’t have to pay for it).
  2. Do all your measurements as three corner hats. You run two references and one DUT into gear that will do that sort of test.
  3. Segment the measurements and use carefully selected references for those ranges.

None of those are actually simple. Number 3 sounds cool until you realize that you are switching test setups around a lot and
the devices you are using still need a setup like 2 to figure out which ones to use.

So do you need a GPS? What are the limits on your MTIE tests? (MTIE on an OCXO is highly dependent on several
things so there is no simple number there). A very normal quartz based GPSDO might be a fine reference for your test.

How much shorter are the other tests? Is ADEV at 1,000 seconds even of interest? If the answer is < 1,000 seconds a
Rb may not do you much good at all.

Lots of twists and turns.

Bob

On Feb 11, 2017, at 5:52 PM, gkk gb modjklist@comcast.net wrote:

Thanks Bob,

I should clarify the MTIE measurement extends 100000 seconds (the others are less time). Is it a reasonable question to ask if GPS is needed? Or are there other variables that are involved?

Good point about the temperature stability, I hadn't considered that. Can I place in a temperature chamber to provide a better thermal environment, or does that cause other issues (vibration from blowers, EMI noise, etc.)? Other ways to mitigate temperature changes?

It seems a Rubidium is good after a timescale of 100 s. What do people do below 100 s to characterize quartz oscillators. Do they simply try to find the most stable parts they can afford and break the x-axis (tau) into two regions using difference references for each? If so, are there generally accepted "gold" standards anyone can recommend for crystal products with the best stability to use as a reference between 0.1 and 100 seconds, for example?

On February 11, 2017 at 6:29 AM Bob Camp kb8tq@n1k.org wrote:

Hi

Backing up a bit here.

On Feb 10, 2017, at 7:35 PM, gkk gb modjklist@comcast.net wrote:

Hello experts, I need a Rubidium frequency reference for my company, and wonder if I also need to GPS discipline it.

I characterize crystal-based OCXOs for ADEV, MTIE, and TDEV, and my longest measurement time is 100,000 seconds (28 hours).

If your longest measurement is a 100,000 second ADEV, then your measurement time will be out in the
1,000,000 to 10,000,000 second range. Is that really what you are doing?

If 100,000 seconds ADEV is your longest measurement, what is the shortest tau you are interested in?
A Rb is not going to be much use for testing a good OCXO at shorter tau. Where the crossover happens
depends a lot on the grade of OCXO you are working with. By the time you get to 1 second
most OCXO’s will be noticeably better than most Rb’s.

I'm looking at this graph from SRS for PRS10,

http://www.thinksrs.com/assets/instr/PRS10/PRS10diag2LG.gif

I would suggest that plot is probably not the best one to depend on for GPS performance. In a GPSDO setting
the cut over points are all over the place depending on which design you look at.

and thinking that as long as I calibrate a Rubidium source annually, there's no need for a GPS (since it only appears to degrade stability). Is this true in general, or is the graph misleading me because it may be true here, but not always.

The big issue is going to be temperature stability. If you have a Rb that is (say) 5x10^-10 over 0 to 50C, that is likely 1x10^-11 / C (or maybe more). A 2C delta in
your lab as the HVAC cycles will give you a 2x10^-11 “hump” in your ADEV plot.

Also consider that if you want an “easy” measurement of the devices you are testing, the reference source probably should be
5X better than what you expect out of the DUT. You probably will not have that luxury in this case. That gets you into multiple
references and things like three corner hat testing.

So my question, is a GPS necessary to discipline a Rubidium standard to characterize the best crystal oscillators for stability, or can I do without it (and just calibrate the Rubidium annually to maintain accuracy) and actually get better stability?

How many seconds out is a GPS generally needed to improve accuracy from a Rubidium standard?

If you really are running 1,000,000 to 10,000,000 second long tests, you need the GPS.

Lots of variables

Bob


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi You likely need to do this with something like a TimePod to get the resolution. Setting up three of them (or anything similar) is going to cost you way more than the Rb. If you have a 100 ps counter, you will get 1x10^-10 at 1 second for ADEV data. That will go to 1x10^-11 at 10 seconds and 1x10^-12 at 100 seconds. You only hit 1x10^-13 at 1,000 seconds. Again it’s a “resolution 5X better than measurement” sort of thing. You don’t want to read the floor of your test gear, you want to see the device under test. If you have a 20 ps counter the numbers above would be your 5X limits rather than the floor. Again, this gets back to the “what do you want to do?” and “how short is your shortest tau?” questions. In the actual use case, there may be no interest at all in Tau’s shorter than 1,000 seconds. It’s also possible that the OCXO’s have an expected floor at 1x10^-10 from 1 second to 1,000 seconds. There are lots of variables. Bob > On Feb 12, 2017, at 7:12 AM, Peter Reilley <preilley_454@comcast.net> wrote: > > Could you do all three tests in parallel? One unit under test driving three counters. > Each counter using a different reference signal, one on a OCXO, one on a rubidium, > and one on a GPS disciplined oscillator. At each point in time during the test simply > choose the one that gives the best ADEV? > > Pete. > > On 2/11/2017 6:52 PM, Bob Camp wrote: >> Hi >> >> Using ADEV as an example (the other stuff will have it’s own curves, but the result is the same): >> >> A typical Rb should have a stability at short tau that goes as 1/ square root(Tau). If you are at 2x10^-11 at 1 second, you >> will be at 2x10^-12 at 100 seconds and 2x10^-13 at 1,000 seconds. Somewhere in the parts in 10^-13 that relation will >> start to diverge from reality. >> >> A fairly normal low frequency OCXO has a stability that is fairly flat with tau in the 1 to 100 second range. If they have been >> on power constantly that “flat zone" may extend to 1,000 seconds. Floors should be in the low parts in 10^-12 to mid parts >> in 10^-13 range. >> >> A good OCXO *may* beat a normal Rb at 1,000 seconds. That may or may not be an issue in your case. It depends a lot >> on what you are trying to do. >> >> Simple solutions: >> >> 1) Run something better than an Rb. A hydrogen maser is one alternative (simple if you don’t have to pay for it). >> 2) Do all your measurements as three corner hats. You run two references and one DUT into gear that will do that sort of test. >> 3) Segment the measurements and use carefully selected references for those ranges. >> >> None of those are actually simple. Number 3 sounds cool until you realize that you are switching test setups around a lot and >> the devices you are using still need a setup like 2 to figure out which ones to use. >> >> So do you need a GPS? What are the limits on your MTIE tests? (MTIE on an OCXO is highly dependent on several >> things so there is no simple number there). A very normal quartz based GPSDO might be a fine reference for your test. >> >> How much shorter are the other tests? Is ADEV at 1,000 seconds even of interest? If the answer is < 1,000 seconds a >> Rb may not do you much good at all. >> >> Lots of twists and turns. >> >> Bob >> >> >> >> >> >>> On Feb 11, 2017, at 5:52 PM, gkk gb <modjklist@comcast.net> wrote: >>> >>> Thanks Bob, >>> >>> >>> >>> I should clarify the MTIE measurement extends 100000 seconds (the others are less time). Is it a reasonable question to ask if GPS is needed? Or are there other variables that are involved? >>> >>> >>> >>> Good point about the temperature stability, I hadn't considered that. Can I place in a temperature chamber to provide a better thermal environment, or does that cause other issues (vibration from blowers, EMI noise, etc.)? Other ways to mitigate temperature changes? >>> >>> >>> >>> It seems a Rubidium is good after a timescale of 100 s. What do people do below 100 s to characterize quartz oscillators. Do they simply try to find the most stable parts they can afford and break the x-axis (tau) into two regions using difference references for each? If so, are there generally accepted "gold" standards anyone can recommend for crystal products with the best stability to use as a reference between 0.1 and 100 seconds, for example? >>> >>> >>> >>> On February 11, 2017 at 6:29 AM Bob Camp <kb8tq@n1k.org> wrote: >>> >>> Hi >>> >>> Backing up a bit here. >>> >>> On Feb 10, 2017, at 7:35 PM, gkk gb <modjklist@comcast.net> wrote: >>> >>> Hello experts, I need a Rubidium frequency reference for my company, and wonder if I also need to GPS discipline it. >>> >>> I characterize crystal-based OCXOs for ADEV, MTIE, and TDEV, and my longest measurement time is 100,000 seconds (28 hours). >>> >>> If your longest measurement is a 100,000 second ADEV, then your measurement time will be out in the >>> 1,000,000 to 10,000,000 second range. Is that really what you are doing? >>> >>> If 100,000 seconds ADEV is your longest measurement, what is the shortest tau you are interested in? >>> A Rb is not going to be much use for testing a good OCXO at shorter tau. Where the crossover happens >>> depends a lot on the grade of OCXO you are working with. By the time you get to 1 second >>> most OCXO’s will be noticeably better than most Rb’s. >>> >>> I'm looking at this graph from SRS for PRS10, >>> >>> http://www.thinksrs.com/assets/instr/PRS10/PRS10diag2LG.gif >>> >>> I would suggest that plot is probably not the best one to depend on for GPS performance. In a GPSDO setting >>> the cut over points are all over the place depending on which design you look at. >>> >>> and thinking that as long as I calibrate a Rubidium source annually, there's no need for a GPS (since it only appears to degrade stability). Is this true in general, or is the graph misleading me because it may be true here, but not always. >>> >>> The big issue is going to be temperature stability. If you have a Rb that is (say) 5x10^-10 over 0 to 50C, that is likely 1x10^-11 / C (or maybe more). A 2C delta in >>> your lab as the HVAC cycles will give you a 2x10^-11 “hump” in your ADEV plot. >>> >>> Also consider that if you want an “easy” measurement of the devices you are testing, the reference source probably should be >>> 5X better than what you expect out of the DUT. You probably will not have that luxury in this case. That gets you into multiple >>> references and things like three corner hat testing. >>> >>> So my question, is a GPS necessary to discipline a Rubidium standard to characterize the best crystal oscillators for stability, or can I do without it (and just calibrate the Rubidium annually to maintain accuracy) and actually get better stability? >>> >>> How many seconds out is a GPS generally needed to improve accuracy from a Rubidium standard? >>> >>> If you really are running 1,000,000 to 10,000,000 second long tests, you need the GPS. >>> >>> Lots of variables >>> >>> Bob >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BC
Bob Camp
Sun, Feb 12, 2017 9:16 PM

Hi

How an ADEV floor do you expect to see on your devices?

How short a Tau are you after.

Depending on the answers to those questions, splitting things can be pretty easy
or pretty hard. Isolating signals between multiple test sets is a bit of a pain if the
splitting is not done with active devices.

The phase data used for all of your tests is identical. The test run for one is no
different than the test run for any of the rest of them. The process consists of
obtaining a data set and then running the appropriate statistical evaluation
on that data.

Bob

On Feb 12, 2017, at 1:16 PM, gkk gb modjklist@comcast.net wrote:

That is an interesting suggestion (thanks), and would indeed work for me if it is possible to split the DUT signal into 3 signals in a such a way that wouldn't affect measurements for ADEV, TDEV, MTIE. But I'm thinking anything active would introduce it's own noise into the signal and change the data. Has this technique been used to success in the past?

On February 12, 2017 at 4:12 AM Peter Reilley preilley_454@comcast.net wrote:

Could you do all three tests in parallel?  One unit under test driving
three counters.
Each counter using a different reference signal, one on a OCXO, one on a
rubidium,
and one on a GPS disciplined oscillator.  At each point in time during
the test simply
choose the one that gives the best ADEV?

Pete.

On 2/11/2017 6:52 PM, Bob Camp wrote:

Hi

Using ADEV as an example (the other stuff will have it’s own curves, but the result is the same):

A typical Rb should have a stability at short tau that goes as 1/ square root(Tau). If you are at 2x10^-11 at 1 second, you
will be at 2x10^-12 at 100 seconds and 2x10^-13 at 1,000 seconds. Somewhere in the parts in 10^-13 that relation will
start to diverge from reality.

A fairly normal low frequency OCXO has a stability that is fairly flat with tau in the 1 to 100 second range. If they have been
on power constantly that “flat zone" may extend to 1,000 seconds. Floors should be in the low parts in 10^-12 to mid parts
in 10^-13 range.

A good OCXO may beat a normal Rb at 1,000 seconds. That may or may not be an issue in your case. It depends a lot
on what you are trying to do.

Simple solutions:

  1. Run something better than an Rb. A hydrogen maser is one alternative (simple if you don’t have to pay for it).
  2. Do all your measurements as three corner hats. You run two references and one DUT into gear that will do that sort of test.
  3. Segment the measurements and use carefully selected references for those ranges.

None of those are actually simple. Number 3 sounds cool until you realize that you are switching test setups around a lot and
the devices you are using still need a setup like 2 to figure out which ones to use.

So do you need a GPS? What are the limits on your MTIE tests? (MTIE on an OCXO is highly dependent on several
things so there is no simple number there). A very normal quartz based GPSDO might be a fine reference for your test.

How much shorter are the other tests? Is ADEV at 1,000 seconds even of interest? If the answer is < 1,000 seconds a
Rb may not do you much good at all.

Lots of twists and turns.

Bob

On Feb 11, 2017, at 5:52 PM, gkk gb modjklist@comcast.net wrote:

Thanks Bob,

I should clarify the MTIE measurement extends 100000 seconds (the others are less time). Is it a reasonable question to ask if GPS is needed? Or are there other variables that are involved?

Good point about the temperature stability, I hadn't considered that. Can I place in a temperature chamber to provide a better thermal environment, or does that cause other issues (vibration from blowers, EMI noise, etc.)? Other ways to mitigate temperature changes?

It seems a Rubidium is good after a timescale of 100 s. What do people do below 100 s to characterize quartz oscillators. Do they simply try to find the most stable parts they can afford and break the x-axis (tau) into two regions using difference references for each? If so, are there generally accepted "gold" standards anyone can recommend for crystal products with the best stability to use as a reference between 0.1 and 100 seconds, for example?

On February 11, 2017 at 6:29 AM Bob Camp kb8tq@n1k.org wrote:

Hi

Backing up a bit here.

On Feb 10, 2017, at 7:35 PM, gkk gb modjklist@comcast.net wrote:

Hello experts, I need a Rubidium frequency reference for my company, and wonder if I also need to GPS discipline it.

I characterize crystal-based OCXOs for ADEV, MTIE, and TDEV, and my longest measurement time is 100,000 seconds (28 hours).

If your longest measurement is a 100,000 second ADEV, then your measurement time will be out in the
1,000,000 to 10,000,000 second range. Is that really what you are doing?

If 100,000 seconds ADEV is your longest measurement, what is the shortest tau you are interested in?
A Rb is not going to be much use for testing a good OCXO at shorter tau. Where the crossover happens
depends a lot on the grade of OCXO you are working with. By the time you get to 1 second
most OCXO’s will be noticeably better than most Rb’s.

I'm looking at this graph from SRS for PRS10,

http://www.thinksrs.com/assets/instr/PRS10/PRS10diag2LG.gif

I would suggest that plot is probably not the best one to depend on for GPS performance. In a GPSDO setting
the cut over points are all over the place depending on which design you look at.

and thinking that as long as I calibrate a Rubidium source annually, there's no need for a GPS (since it only appears to degrade stability). Is this true in general, or is the graph misleading me because it may be true here, but not always.

The big issue is going to be temperature stability. If you have a Rb that is (say) 5x10^-10 over 0 to 50C, that is likely 1x10^-11 / C (or maybe more). A 2C delta in
your lab as the HVAC cycles will give you a 2x10^-11 “hump” in your ADEV plot.

Also consider that if you want an “easy” measurement of the devices you are testing, the reference source probably should be
5X better than what you expect out of the DUT. You probably will not have that luxury in this case. That gets you into multiple
references and things like three corner hat testing.

So my question, is a GPS necessary to discipline a Rubidium standard to characterize the best crystal oscillators for stability, or can I do without it (and just calibrate the Rubidium annually to maintain accuracy) and actually get better stability?

How many seconds out is a GPS generally needed to improve accuracy from a Rubidium standard?

If you really are running 1,000,000 to 10,000,000 second long tests, you need the GPS.

Lots of variables

Bob


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi How an ADEV floor do you expect to see on your devices? How short a Tau are you after. Depending on the answers to those questions, splitting things can be pretty easy or pretty hard. Isolating signals between multiple test sets is a bit of a pain if the splitting is not done with active devices. The phase data used for all of your tests is identical. The test run for one is no different than the test run for any of the rest of them. The process consists of obtaining a data set and then running the appropriate statistical evaluation on that data. Bob > On Feb 12, 2017, at 1:16 PM, gkk gb <modjklist@comcast.net> wrote: > > That is an interesting suggestion (thanks), and would indeed work for me if it is possible to split the DUT signal into 3 signals in a such a way that wouldn't affect measurements for ADEV, TDEV, MTIE. But I'm thinking anything active would introduce it's own noise into the signal and change the data. Has this technique been used to success in the past? > >> On February 12, 2017 at 4:12 AM Peter Reilley <preilley_454@comcast.net> wrote: >> >> >> Could you do all three tests in parallel? One unit under test driving >> three counters. >> Each counter using a different reference signal, one on a OCXO, one on a >> rubidium, >> and one on a GPS disciplined oscillator. At each point in time during >> the test simply >> choose the one that gives the best ADEV? >> >> Pete. >> >> On 2/11/2017 6:52 PM, Bob Camp wrote: >>> Hi >>> >>> Using ADEV as an example (the other stuff will have it’s own curves, but the result is the same): >>> >>> A typical Rb should have a stability at short tau that goes as 1/ square root(Tau). If you are at 2x10^-11 at 1 second, you >>> will be at 2x10^-12 at 100 seconds and 2x10^-13 at 1,000 seconds. Somewhere in the parts in 10^-13 that relation will >>> start to diverge from reality. >>> >>> A fairly normal low frequency OCXO has a stability that is fairly flat with tau in the 1 to 100 second range. If they have been >>> on power constantly that “flat zone" may extend to 1,000 seconds. Floors should be in the low parts in 10^-12 to mid parts >>> in 10^-13 range. >>> >>> A good OCXO *may* beat a normal Rb at 1,000 seconds. That may or may not be an issue in your case. It depends a lot >>> on what you are trying to do. >>> >>> Simple solutions: >>> >>> 1) Run something better than an Rb. A hydrogen maser is one alternative (simple if you don’t have to pay for it). >>> 2) Do all your measurements as three corner hats. You run two references and one DUT into gear that will do that sort of test. >>> 3) Segment the measurements and use carefully selected references for those ranges. >>> >>> None of those are actually simple. Number 3 sounds cool until you realize that you are switching test setups around a lot and >>> the devices you are using still need a setup like 2 to figure out which ones to use. >>> >>> So do you need a GPS? What are the limits on your MTIE tests? (MTIE on an OCXO is highly dependent on several >>> things so there is no simple number there). A very normal quartz based GPSDO might be a fine reference for your test. >>> >>> How much shorter are the other tests? Is ADEV at 1,000 seconds even of interest? If the answer is < 1,000 seconds a >>> Rb may not do you much good at all. >>> >>> Lots of twists and turns. >>> >>> Bob >>> >>> >>> >>> >>> >>>> On Feb 11, 2017, at 5:52 PM, gkk gb <modjklist@comcast.net> wrote: >>>> >>>> Thanks Bob, >>>> >>>> >>>> >>>> I should clarify the MTIE measurement extends 100000 seconds (the others are less time). Is it a reasonable question to ask if GPS is needed? Or are there other variables that are involved? >>>> >>>> >>>> >>>> Good point about the temperature stability, I hadn't considered that. Can I place in a temperature chamber to provide a better thermal environment, or does that cause other issues (vibration from blowers, EMI noise, etc.)? Other ways to mitigate temperature changes? >>>> >>>> >>>> >>>> It seems a Rubidium is good after a timescale of 100 s. What do people do below 100 s to characterize quartz oscillators. Do they simply try to find the most stable parts they can afford and break the x-axis (tau) into two regions using difference references for each? If so, are there generally accepted "gold" standards anyone can recommend for crystal products with the best stability to use as a reference between 0.1 and 100 seconds, for example? >>>> >>>> >>>> >>>> On February 11, 2017 at 6:29 AM Bob Camp <kb8tq@n1k.org> wrote: >>>> >>>> Hi >>>> >>>> Backing up a bit here. >>>> >>>> On Feb 10, 2017, at 7:35 PM, gkk gb <modjklist@comcast.net> wrote: >>>> >>>> Hello experts, I need a Rubidium frequency reference for my company, and wonder if I also need to GPS discipline it. >>>> >>>> I characterize crystal-based OCXOs for ADEV, MTIE, and TDEV, and my longest measurement time is 100,000 seconds (28 hours). >>>> >>>> If your longest measurement is a 100,000 second ADEV, then your measurement time will be out in the >>>> 1,000,000 to 10,000,000 second range. Is that really what you are doing? >>>> >>>> If 100,000 seconds ADEV is your longest measurement, what is the shortest tau you are interested in? >>>> A Rb is not going to be much use for testing a good OCXO at shorter tau. Where the crossover happens >>>> depends a lot on the grade of OCXO you are working with. By the time you get to 1 second >>>> most OCXO’s will be noticeably better than most Rb’s. >>>> >>>> I'm looking at this graph from SRS for PRS10, >>>> >>>> http://www.thinksrs.com/assets/instr/PRS10/PRS10diag2LG.gif >>>> >>>> I would suggest that plot is probably not the best one to depend on for GPS performance. In a GPSDO setting >>>> the cut over points are all over the place depending on which design you look at. >>>> >>>> and thinking that as long as I calibrate a Rubidium source annually, there's no need for a GPS (since it only appears to degrade stability). Is this true in general, or is the graph misleading me because it may be true here, but not always. >>>> >>>> The big issue is going to be temperature stability. If you have a Rb that is (say) 5x10^-10 over 0 to 50C, that is likely 1x10^-11 / C (or maybe more). A 2C delta in >>>> your lab as the HVAC cycles will give you a 2x10^-11 “hump” in your ADEV plot. >>>> >>>> Also consider that if you want an “easy” measurement of the devices you are testing, the reference source probably should be >>>> 5X better than what you expect out of the DUT. You probably will not have that luxury in this case. That gets you into multiple >>>> references and things like three corner hat testing. >>>> >>>> So my question, is a GPS necessary to discipline a Rubidium standard to characterize the best crystal oscillators for stability, or can I do without it (and just calibrate the Rubidium annually to maintain accuracy) and actually get better stability? >>>> >>>> How many seconds out is a GPS generally needed to improve accuracy from a Rubidium standard? >>>> >>>> If you really are running 1,000,000 to 10,000,000 second long tests, you need the GPS. >>>> >>>> Lots of variables >>>> >>>> Bob >>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>> and follow the instructions there. >>>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
PV
Peter Vince
Mon, Feb 13, 2017 11:40 AM

Surely you just loop it through two counters using a T-piece, with those
counters set to high impedance input, then only terminate the final one?

 Peter

On 12 February 2017 at 18:16, gkk gb modjklist@comcast.net wrote:

That is an interesting suggestion (thanks), and would indeed work for me
if it is possible to split the DUT signal into 3 signals in a such a way
that wouldn't affect measurements for ADEV, TDEV, MTIE. But I'm thinking
anything active would introduce it's own noise into the signal and change
the data. Has this technique been used to success in the past?

Surely you just loop it through two counters using a T-piece, with those counters set to high impedance input, then only terminate the final one? Peter On 12 February 2017 at 18:16, gkk gb <modjklist@comcast.net> wrote: > That is an interesting suggestion (thanks), and would indeed work for me > if it is possible to split the DUT signal into 3 signals in a such a way > that wouldn't affect measurements for ADEV, TDEV, MTIE. But I'm thinking > anything active would introduce it's own noise into the signal and change > the data. Has this technique been used to success in the past? > >