time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

Re: [time-nuts] Papers on timing for lunar laser ranging

E
EWKehren@aol.com
Sun, Jul 16, 2017 12:21 PM

As part of our GPSDO work Richard Mc Corkle and I tested multiple DAC's
using hardware I developed and Richard wrote the code. We limited to
affordable  and solderable. The LTC1655 was the clear winner because of linearity and
temperature, see attached. We tested dithering 20 bits and stacking two
for  coarse fine, storing the test data from using the LTC2400 ADC. Limited
our  choices to dither and bare use. We did this 5 years ago. Contact with
Richard  has sadly stopped. I am very concerned, however we continue to use his
contributions on several projects with very good results. We use the LTC
1655 on  Rb's because its resolution and range as is, is perfect.
Five years later I know no better alternative
Bert Kehren

In a message dated 7/16/2017 5:09:08 A.M. Eastern Daylight Time,
attila@kinali.ch writes:

On Sat,  8 Jul 2017 11:42:44 -0700
Tim Lister listertim@gmail.com  wrote:

Forgive the ignorance, but why is there a large disparity  between ADC
and DAC capabilities ?
For example, Linear  Technology sell a 24 bit ADC for ~$7 but an 18 bit
DAC is  $30-50...

Much simplified, it boils down to it being easier to measure  voltage
differences by averaging than keeping a voltage  constant.

E.g. in those >20bit ADC's you will usually find a  delta-sigma ADC,
usually 3rd to 5th order with a 1.5 to 5 bit ADC/DAC  inside. The ADC
and DAC can be laser trimmed to be in the order of 0.1% of  their
ideal values. With a few additional tricks you can get the most  of
the remaining non-linearity out. These tricks also help to  remove
errors due to DC-offsets in the signal path. But the  biggest
improvement comes from averaging over many "samples" to get  the
white noise out. If you look at the usual sample rates at  which
those ADC reach their "full" performance, it is around
1-30  (output) samples per second.

On the other hand, on a DAC you need to  keep the output voltage
stable. You can do the same delta-sigma approach as  with the ADC
with much the same result, but you have one big problem:
it  is not easy to build an analog low pass filter that has a corner
frequency  down at 10Hz. This means, you have to work at a much higher
frequency to  have a low pass filter that can be realized (let's say 1kHz
if you are  building a discrete filter, higher if it's integrated).
But that means that  you have several orders of magnitude more (white)
noise.

Additionally,  a lot of people expect to do a couple of 1000 samples
per second at least,  to have a usefull DAC. But that contradicts the
need to have a narrow band  low pass filter to get the noise out.

Attila Kinali

It is upon moral qualities that a  society is ultimately founded. All
the prosperity and technological  sophistication in the world is of no
use without that  foundation.
-- Miss Matheson, The Diamond Age, Neil  Stephenson


time-nuts  mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions there.

As part of our GPSDO work Richard Mc Corkle and I tested multiple DAC's using hardware I developed and Richard wrote the code. We limited to affordable and solderable. The LTC1655 was the clear winner because of linearity and temperature, see attached. We tested dithering 20 bits and stacking two for coarse fine, storing the test data from using the LTC2400 ADC. Limited our choices to dither and bare use. We did this 5 years ago. Contact with Richard has sadly stopped. I am very concerned, however we continue to use his contributions on several projects with very good results. We use the LTC 1655 on Rb's because its resolution and range as is, is perfect. Five years later I know no better alternative Bert Kehren In a message dated 7/16/2017 5:09:08 A.M. Eastern Daylight Time, attila@kinali.ch writes: On Sat, 8 Jul 2017 11:42:44 -0700 Tim Lister <listertim@gmail.com> wrote: > Forgive the ignorance, but why is there a large disparity between ADC > and DAC capabilities ? > For example, Linear Technology sell a 24 bit ADC for ~$7 but an 18 bit > DAC is $30-50... Much simplified, it boils down to it being easier to measure voltage differences by averaging than keeping a voltage constant. E.g. in those >20bit ADC's you will usually find a delta-sigma ADC, usually 3rd to 5th order with a 1.5 to 5 bit ADC/DAC inside. The ADC and DAC can be laser trimmed to be in the order of 0.1% of their ideal values. With a few additional tricks you can get the most of the remaining non-linearity out. These tricks also help to remove errors due to DC-offsets in the signal path. But the biggest improvement comes from averaging over many "samples" to get the white noise out. If you look at the usual sample rates at which those ADC reach their "full" performance, it is around 1-30 (output) samples per second. On the other hand, on a DAC you need to keep the output voltage stable. You can do the same delta-sigma approach as with the ADC with much the same result, but you have one big problem: it is not easy to build an analog low pass filter that has a corner frequency down at 10Hz. This means, you have to work at a much higher frequency to have a low pass filter that can be realized (let's say 1kHz if you are building a discrete filter, higher if it's integrated). But that means that you have several orders of magnitude more (white) noise. Additionally, a lot of people expect to do a couple of 1000 samples per second at least, to have a usefull DAC. But that contradicts the need to have a narrow band low pass filter to get the noise out. Attila Kinali -- It is upon moral qualities that a society is ultimately founded. All the prosperity and technological sophistication in the world is of no use without that foundation. -- Miss Matheson, The Diamond Age, Neil Stephenson _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
CS
Charles Steinmetz
Sun, Jul 16, 2017 6:24 PM

Bert wrote:

We limited to affordable  and solderable.
The LTC1655 was the clear winner because
of linearity and temperature
*    *    *
Five years later I know no better alternative

Well, to name just the first one that comes to mind, how about the
LTC1650?  Like the 1655, it is available in SO and DIP packages.  Its
differential nonlinearity is >2x better than the 1655, it settles 5x
faster (4uS vs. 20uS to 1 LSB) and is nearly 100x (40dB) quieter
(30nV/sqrtHz vs. 280).  The 1650 has substantially lower glitch energy,
as well (1.8nV-S for the 1650, 12nV-S for the 1655).

Not all of the better specs are required for steering an oscillator at
GPSDO rates, but others are a distinct advantage.

I have not surveyed the field to see what other "SO or easier to solder"
DACs are available with better performance than the 1655, but I'm sure
there are others.

This is to take nothing away from the 1655 -- I have accounted for
thousands of them in my own designs, and think very highly of it.

Best regards,

Charles

Bert wrote: > We limited to affordable and solderable. > The LTC1655 was the clear winner because > of linearity and temperature > * * * > Five years later I know no better alternative Well, to name just the first one that comes to mind, how about the LTC1650? Like the 1655, it is available in SO and DIP packages. Its differential nonlinearity is >2x better than the 1655, it settles 5x faster (4uS vs. 20uS to 1 LSB) and is nearly 100x (40dB) quieter (30nV/sqrtHz vs. 280). The 1650 has substantially lower glitch energy, as well (1.8nV-S for the 1650, 12nV-S for the 1655). Not all of the better specs are required for steering an oscillator at GPSDO rates, but others are a distinct advantage. I have not surveyed the field to see what other "SO or easier to solder" DACs are available with better performance than the 1655, but I'm sure there are others. This is to take nothing away from the 1655 -- I have accounted for thousands of them in my own designs, and think very highly of it. Best regards, Charles
BK
Bob kb8tq
Sun, Jul 16, 2017 8:51 PM

Hi

One gotcha with any ADC or DAC is going to be the reference. There, you are in the same
“get what you pay for” dilemma. Stable and noisy, can do. Quiet and not very stable, can do.
Both stable and quiet, not so easy if you want it cheap.

Noise can also be the sigma delta ADC’s weak point. Even at slow rates, some of them need
a lot of averages to quiet down.

Does this or that design need this or that level of stability or noise? That depends a lot on the
approach used. In a GPSDO, cutting down on the EFC range is a great way to “cost reduce” the
rest of the circuit in terms of noise and stability.  Well made modern OCXO’s don’t drift a lot ….

Bob

On Jul 16, 2017, at 2:24 PM, Charles Steinmetz csteinmetz@yandex.com wrote:

Bert wrote:

We limited to affordable  and solderable.
The LTC1655 was the clear winner because
of linearity and temperature
*    *    *
Five years later I know no better alternative

Well, to name just the first one that comes to mind, how about the LTC1650?  Like the 1655, it is available in SO and DIP packages.  Its differential nonlinearity is >2x better than the 1655, it settles 5x faster (4uS vs. 20uS to 1 LSB) and is nearly 100x (40dB) quieter (30nV/sqrtHz vs. 280).  The 1650 has substantially lower glitch energy, as well (1.8nV-S for the 1650, 12nV-S for the 1655).

Not all of the better specs are required for steering an oscillator at GPSDO rates, but others are a distinct advantage.

I have not surveyed the field to see what other "SO or easier to solder" DACs are available with better performance than the 1655, but I'm sure there are others.

This is to take nothing away from the 1655 -- I have accounted for thousands of them in my own designs, and think very highly of it.

Best regards,

Charles


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi One gotcha with any ADC or DAC is going to be the reference. There, you are in the same “get what you pay for” dilemma. Stable and noisy, can do. Quiet and not very stable, can do. Both stable and quiet, not so easy if you want it cheap. Noise can also be the sigma delta ADC’s weak point. Even at slow rates, some of them need a lot of averages to quiet down. Does this or that design need this or that level of stability or noise? That depends a lot on the approach used. In a GPSDO, cutting down on the EFC range is a great way to “cost reduce” the rest of the circuit in terms of noise and stability. Well made modern OCXO’s don’t drift a lot …. Bob > On Jul 16, 2017, at 2:24 PM, Charles Steinmetz <csteinmetz@yandex.com> wrote: > > Bert wrote: > >> We limited to affordable and solderable. >> The LTC1655 was the clear winner because >> of linearity and temperature >> * * * >> Five years later I know no better alternative > > Well, to name just the first one that comes to mind, how about the LTC1650? Like the 1655, it is available in SO and DIP packages. Its differential nonlinearity is >2x better than the 1655, it settles 5x faster (4uS vs. 20uS to 1 LSB) and is nearly 100x (40dB) quieter (30nV/sqrtHz vs. 280). The 1650 has substantially lower glitch energy, as well (1.8nV-S for the 1650, 12nV-S for the 1655). > > Not all of the better specs are required for steering an oscillator at GPSDO rates, but others are a distinct advantage. > > I have not surveyed the field to see what other "SO or easier to solder" DACs are available with better performance than the 1655, but I'm sure there are others. > > This is to take nothing away from the 1655 -- I have accounted for thousands of them in my own designs, and think very highly of it. > > Best regards, > > Charles > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
R(
Richard (Rick) Karlquist
Sun, Jul 16, 2017 10:33 PM

On 7/16/2017 1:51 PM, Bob kb8tq wrote:

Hi

One gotcha with any ADC or DAC is going to be the reference. There, you are in the same
“get what you pay for” dilemma. Stable and noisy, can do. Quiet and not very stable, can do.
Both stable and quiet, not so easy if you want it cheap.

Noise can also be the sigma delta ADC’s weak point. Even at slow rates, some of them need
a lot of averages to quiet down.

The reference initially used in the E1938A turned out to be too
noisy/unstable.  It was non trivial to find an upgrade.  The
HP Smart Clocks of 20 years ago were limited in their performance
by the reference used.

Has there been much improvement in references in the intervening
20 years?

Rick N6RK

On 7/16/2017 1:51 PM, Bob kb8tq wrote: > Hi > > One gotcha with any ADC or DAC is going to be the reference. There, you are in the same > “get what you pay for” dilemma. Stable and noisy, can do. Quiet and not very stable, can do. > Both stable and quiet, not so easy if you want it cheap. > > Noise can also be the sigma delta ADC’s weak point. Even at slow rates, some of them need > a lot of averages to quiet down. > The reference initially used in the E1938A turned out to be too noisy/unstable. It was non trivial to find an upgrade. The HP Smart Clocks of 20 years ago were limited in their performance by the reference used. Has there been much improvement in references in the intervening 20 years? Rick N6RK
BK
Bob kb8tq
Sun, Jul 16, 2017 11:10 PM

Hi

On Jul 16, 2017, at 6:33 PM, Richard (Rick) Karlquist richard@karlquist.com wrote:

On 7/16/2017 1:51 PM, Bob kb8tq wrote:

Hi
One gotcha with any ADC or DAC is going to be the reference. There, you are in the same
“get what you pay for” dilemma. Stable and noisy, can do. Quiet and not very stable, can do.
Both stable and quiet, not so easy if you want it cheap.
Noise can also be the sigma delta ADC’s weak point. Even at slow rates, some of them need
a lot of averages to quiet down.

The reference initially used in the E1938A turned out to be too noisy/unstable.  It was non trivial to find an upgrade.  The
HP Smart Clocks of 20 years ago were limited in their performance
by the reference used.

Has there been much improvement in references in the intervening
20 years?

They still don’t seem to have the hysteresis problem licked. Yes, you can do an oversized reference
and take care of the issue. More or less that’s what you would have done 20 years ago.

Bob

Rick N6RK

Hi > On Jul 16, 2017, at 6:33 PM, Richard (Rick) Karlquist <richard@karlquist.com> wrote: > > > > On 7/16/2017 1:51 PM, Bob kb8tq wrote: >> Hi >> One gotcha with any ADC or DAC is going to be the reference. There, you are in the same >> “get what you pay for” dilemma. Stable and noisy, can do. Quiet and not very stable, can do. >> Both stable and quiet, not so easy if you want it cheap. >> Noise can also be the sigma delta ADC’s weak point. Even at slow rates, some of them need >> a lot of averages to quiet down. > > The reference initially used in the E1938A turned out to be too noisy/unstable. It was non trivial to find an upgrade. The > HP Smart Clocks of 20 years ago were limited in their performance > by the reference used. > > Has there been much improvement in references in the intervening > 20 years? They still don’t seem to have the hysteresis problem licked. Yes, you can do an oversized reference and take care of the issue. More or less that’s what you would have done 20 years ago. Bob > > Rick N6RK
CA
Chris Albertson
Sun, Jul 16, 2017 11:25 PM

What about josephson standards?  After all, this is "Time Nuts" and we are
allowed to propose silly-complex solutions to simple problems if it
improves performance even a little.

But seriously I thought the issue of making a perfect voltage standard was
solved because the Volt is defined to be whatever the Josephson array
produces.    Yes expensive because to runs at nearly absolute zero.

On Sun, Jul 16, 2017 at 4:10 PM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

On Jul 16, 2017, at 6:33 PM, Richard (Rick) Karlquist <

On 7/16/2017 1:51 PM, Bob kb8tq wrote:

Hi
One gotcha with any ADC or DAC is going to be the reference. There, you

are in the same

“get what you pay for” dilemma. Stable and noisy, can do. Quiet and not

very stable, can do.

Both stable and quiet, not so easy if you want it cheap.
Noise can also be the sigma delta ADC’s weak point. Even at slow rates,

some of them need

a lot of averages to quiet down.

The reference initially used in the E1938A turned out to be too

noisy/unstable.  It was non trivial to find an upgrade.  The

HP Smart Clocks of 20 years ago were limited in their performance
by the reference used.

Has there been much improvement in references in the intervening
20 years?

They still don’t seem to have the hysteresis problem licked. Yes, you can
do an oversized reference
and take care of the issue. More or less that’s what you would have done
20 years ago.

Bob

Rick N6RK


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

--

Chris Albertson
Redondo Beach, California

What about josephson standards? After all, this is "Time Nuts" and we are allowed to propose silly-complex solutions to simple problems if it improves performance even a little. But seriously I thought the issue of making a perfect voltage standard was solved because the Volt is defined to be whatever the Josephson array produces. Yes expensive because to runs at nearly absolute zero. On Sun, Jul 16, 2017 at 4:10 PM, Bob kb8tq <kb8tq@n1k.org> wrote: > Hi > > > On Jul 16, 2017, at 6:33 PM, Richard (Rick) Karlquist < > richard@karlquist.com> wrote: > > > > > > > > On 7/16/2017 1:51 PM, Bob kb8tq wrote: > >> Hi > >> One gotcha with any ADC or DAC is going to be the reference. There, you > are in the same > >> “get what you pay for” dilemma. Stable and noisy, can do. Quiet and not > very stable, can do. > >> Both stable and quiet, not so easy if you want it cheap. > >> Noise can also be the sigma delta ADC’s weak point. Even at slow rates, > some of them need > >> a lot of averages to quiet down. > > > > The reference initially used in the E1938A turned out to be too > noisy/unstable. It was non trivial to find an upgrade. The > > HP Smart Clocks of 20 years ago were limited in their performance > > by the reference used. > > > > Has there been much improvement in references in the intervening > > 20 years? > > They still don’t seem to have the hysteresis problem licked. Yes, you can > do an oversized reference > and take care of the issue. More or less that’s what you would have done > 20 years ago. > > Bob > > > > > > Rick N6RK > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > and follow the instructions there. > -- Chris Albertson Redondo Beach, California
BK
Bob kb8tq
Mon, Jul 17, 2017 1:49 AM

HI

This is a limitation on an OCXO based GPSDO. That’s really the bottom line here.
It’s a limitation in an OCXO based part, but not in one based on an Rb or a Cs. If
the added component costs far more than a Cs, it’s not an answer.

Bob

On Jul 16, 2017, at 7:25 PM, Chris Albertson albertson.chris@gmail.com wrote:

What about josephson standards?  After all, this is "Time Nuts" and we are
allowed to propose silly-complex solutions to simple problems if it
improves performance even a little.

But seriously I thought the issue of making a perfect voltage standard was
solved because the Volt is defined to be whatever the Josephson array
produces.    Yes expensive because to runs at nearly absolute zero.

On Sun, Jul 16, 2017 at 4:10 PM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

On Jul 16, 2017, at 6:33 PM, Richard (Rick) Karlquist <

On 7/16/2017 1:51 PM, Bob kb8tq wrote:

Hi
One gotcha with any ADC or DAC is going to be the reference. There, you

are in the same

“get what you pay for” dilemma. Stable and noisy, can do. Quiet and not

very stable, can do.

Both stable and quiet, not so easy if you want it cheap.
Noise can also be the sigma delta ADC’s weak point. Even at slow rates,

some of them need

a lot of averages to quiet down.

The reference initially used in the E1938A turned out to be too

noisy/unstable.  It was non trivial to find an upgrade.  The

HP Smart Clocks of 20 years ago were limited in their performance
by the reference used.

Has there been much improvement in references in the intervening
20 years?

They still don’t seem to have the hysteresis problem licked. Yes, you can
do an oversized reference
and take care of the issue. More or less that’s what you would have done
20 years ago.

Bob

Rick N6RK


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

--

Chris Albertson
Redondo Beach, California


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

HI This is a limitation on an OCXO based GPSDO. That’s really the bottom line here. It’s a limitation in an OCXO based part, but not in one based on an Rb or a Cs. If the added component costs far more than a Cs, it’s not an answer. Bob > On Jul 16, 2017, at 7:25 PM, Chris Albertson <albertson.chris@gmail.com> wrote: > > What about josephson standards? After all, this is "Time Nuts" and we are > allowed to propose silly-complex solutions to simple problems if it > improves performance even a little. > > But seriously I thought the issue of making a perfect voltage standard was > solved because the Volt is defined to be whatever the Josephson array > produces. Yes expensive because to runs at nearly absolute zero. > > On Sun, Jul 16, 2017 at 4:10 PM, Bob kb8tq <kb8tq@n1k.org> wrote: > >> Hi >> >>> On Jul 16, 2017, at 6:33 PM, Richard (Rick) Karlquist < >> richard@karlquist.com> wrote: >>> >>> >>> >>> On 7/16/2017 1:51 PM, Bob kb8tq wrote: >>>> Hi >>>> One gotcha with any ADC or DAC is going to be the reference. There, you >> are in the same >>>> “get what you pay for” dilemma. Stable and noisy, can do. Quiet and not >> very stable, can do. >>>> Both stable and quiet, not so easy if you want it cheap. >>>> Noise can also be the sigma delta ADC’s weak point. Even at slow rates, >> some of them need >>>> a lot of averages to quiet down. >>> >>> The reference initially used in the E1938A turned out to be too >> noisy/unstable. It was non trivial to find an upgrade. The >>> HP Smart Clocks of 20 years ago were limited in their performance >>> by the reference used. >>> >>> Has there been much improvement in references in the intervening >>> 20 years? >> >> They still don’t seem to have the hysteresis problem licked. Yes, you can >> do an oversized reference >> and take care of the issue. More or less that’s what you would have done >> 20 years ago. >> >> Bob >> >> >>> >>> Rick N6RK >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/ >> mailman/listinfo/time-nuts >> and follow the instructions there. >> > > > > -- > > Chris Albertson > Redondo Beach, California > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
AB
Azelio Boriani
Mon, Jul 17, 2017 5:54 PM

This implies that in a Rb or Cs there is not a voltage reference source?

On Mon, Jul 17, 2017 at 3:49 AM, Bob kb8tq kb8tq@n1k.org wrote:

HI

This is a limitation on an OCXO based GPSDO. That’s really the bottom line here.
It’s a limitation in an OCXO based part, but not in one based on an Rb or a Cs. If
the added component costs far more than a Cs, it’s not an answer.

Bob

On Jul 16, 2017, at 7:25 PM, Chris Albertson albertson.chris@gmail.com wrote:

What about josephson standards?  After all, this is "Time Nuts" and we are
allowed to propose silly-complex solutions to simple problems if it
improves performance even a little.

But seriously I thought the issue of making a perfect voltage standard was
solved because the Volt is defined to be whatever the Josephson array
produces.    Yes expensive because to runs at nearly absolute zero.

On Sun, Jul 16, 2017 at 4:10 PM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

On Jul 16, 2017, at 6:33 PM, Richard (Rick) Karlquist <

On 7/16/2017 1:51 PM, Bob kb8tq wrote:

Hi
One gotcha with any ADC or DAC is going to be the reference. There, you

are in the same

“get what you pay for” dilemma. Stable and noisy, can do. Quiet and not

very stable, can do.

Both stable and quiet, not so easy if you want it cheap.
Noise can also be the sigma delta ADC’s weak point. Even at slow rates,

some of them need

a lot of averages to quiet down.

The reference initially used in the E1938A turned out to be too

noisy/unstable.  It was non trivial to find an upgrade.  The

HP Smart Clocks of 20 years ago were limited in their performance
by the reference used.

Has there been much improvement in references in the intervening
20 years?

They still don’t seem to have the hysteresis problem licked. Yes, you can
do an oversized reference
and take care of the issue. More or less that’s what you would have done
20 years ago.

Bob

Rick N6RK


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

--

Chris Albertson
Redondo Beach, California


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

This implies that in a Rb or Cs there is not a voltage reference source? On Mon, Jul 17, 2017 at 3:49 AM, Bob kb8tq <kb8tq@n1k.org> wrote: > HI > > This is a limitation on an OCXO based GPSDO. That’s really the bottom line here. > It’s a limitation in an OCXO based part, but not in one based on an Rb or a Cs. If > the added component costs far more than a Cs, it’s not an answer. > > Bob > >> On Jul 16, 2017, at 7:25 PM, Chris Albertson <albertson.chris@gmail.com> wrote: >> >> What about josephson standards? After all, this is "Time Nuts" and we are >> allowed to propose silly-complex solutions to simple problems if it >> improves performance even a little. >> >> But seriously I thought the issue of making a perfect voltage standard was >> solved because the Volt is defined to be whatever the Josephson array >> produces. Yes expensive because to runs at nearly absolute zero. >> >> On Sun, Jul 16, 2017 at 4:10 PM, Bob kb8tq <kb8tq@n1k.org> wrote: >> >>> Hi >>> >>>> On Jul 16, 2017, at 6:33 PM, Richard (Rick) Karlquist < >>> richard@karlquist.com> wrote: >>>> >>>> >>>> >>>> On 7/16/2017 1:51 PM, Bob kb8tq wrote: >>>>> Hi >>>>> One gotcha with any ADC or DAC is going to be the reference. There, you >>> are in the same >>>>> “get what you pay for” dilemma. Stable and noisy, can do. Quiet and not >>> very stable, can do. >>>>> Both stable and quiet, not so easy if you want it cheap. >>>>> Noise can also be the sigma delta ADC’s weak point. Even at slow rates, >>> some of them need >>>>> a lot of averages to quiet down. >>>> >>>> The reference initially used in the E1938A turned out to be too >>> noisy/unstable. It was non trivial to find an upgrade. The >>>> HP Smart Clocks of 20 years ago were limited in their performance >>>> by the reference used. >>>> >>>> Has there been much improvement in references in the intervening >>>> 20 years? >>> >>> They still don’t seem to have the hysteresis problem licked. Yes, you can >>> do an oversized reference >>> and take care of the issue. More or less that’s what you would have done >>> 20 years ago. >>> >>> Bob >>> >>> >>>> >>>> Rick N6RK >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>> mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >> >> >> >> -- >> >> Chris Albertson >> Redondo Beach, California >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
R(
Richard (Rick) Karlquist
Mon, Jul 17, 2017 7:07 PM

On 7/17/2017 10:54 AM, Azelio Boriani wrote:

This implies that in a Rb or Cs there is not a voltage reference source?

Yes, that's right, there is no voltage reference with a material
effect on stability or accuracy.

Rick N6RK

On 7/17/2017 10:54 AM, Azelio Boriani wrote: > This implies that in a Rb or Cs there is not a voltage reference source? > Yes, that's right, there is no voltage reference with a material effect on stability or accuracy. Rick N6RK
AK
Attila Kinali
Mon, Jul 17, 2017 8:41 PM

On Mon, 17 Jul 2017 12:07:29 -0700
"Richard (Rick) Karlquist" richard@karlquist.com wrote:

On 7/17/2017 10:54 AM, Azelio Boriani wrote:

This implies that in a Rb or Cs there is not a voltage reference source?

Yes, that's right, there is no voltage reference with a material
effect on stability or accuracy.

How about the temperature and C-field control? Aren't those based
on keeping the output of a sensor stable compared to a voltage reference?

		Attila Kinali

--
You know, the very powerful and the very stupid have one thing in common.
They don't alters their views to fit the facts, they alter the facts to
fit the views, which can be uncomfortable if you happen to be one of the
facts that needs altering.  -- The Doctor

On Mon, 17 Jul 2017 12:07:29 -0700 "Richard (Rick) Karlquist" <richard@karlquist.com> wrote: > On 7/17/2017 10:54 AM, Azelio Boriani wrote: > > This implies that in a Rb or Cs there is not a voltage reference source? > > > > Yes, that's right, there is no voltage reference with a material > effect on stability or accuracy. How about the temperature and C-field control? Aren't those based on keeping the output of a sensor stable compared to a voltage reference? Attila Kinali -- You know, the very powerful and the very stupid have one thing in common. They don't alters their views to fit the facts, they alter the facts to fit the views, which can be uncomfortable if you happen to be one of the facts that needs altering. -- The Doctor