time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

Z3801A OCXO manual trimming

TC
Tom Curlee
Sat, Mar 3, 2018 6:37 PM

Since the Z3801A is being discussed, I thought I'd ask about an issue I'm having with my unit.  I use my Z3801 as my working lab standard for the usual pieces of RF test equipment.  In the past year or so I've had the unit drop out of lock and go into standby mode.  Resetting/cycling power would bring it back into lock for a while, but it generally got worse and now stays in hold over mode.  LH (thanks Mark Sims!!) reports that everything is operating normally except that the it has a PLL unlock.  The one highly suspicious item is that the DAC is at 99.996902% - full output.
The unit shows it has over 94.5K hours run time, so I suspect that the OCXO has aged to the point that the EFC can't pull it into lock.  I disassembled the OCXO to see if it had a trimmer capacitor like the standard 10811 units.  After removing the outer case and foam insulation, I see that the outer heater is one of the thin printed circuit serpentine heaters on what I think is Kapton.  That would need to be peeled off of the case to either get to the hole for the trimmer (if there is one) or to remove the cover for the inner case.
What I don't understand is the purpose of what looks like another coil or heater wrapped around the Kapton printed circuit heater stuck to the inner case.  This second coil/heater is 2 layers of 1/8" thick red foam wrapped completely around the inner case, with fine copper wires wrapped over each layer.  At least I think there are wires on each layer.  This whole second heater is taped down and I don't want to dig any further until I know a bit more about what I'm doing or find that there isn't a trimmer on the double oven 10811 oscillators.
Any ideas on the purpose of the outer heater (or whatever it is)?  Does the double oven 10811 have a manual trimmer that I can adjust to bring the oscillator back to the center of the EFC range?  Assuming that I can manually adjust the OCXO back into adjustment range, will there be any issues with the Z3801 performance, things like phase noise, short term stability, etc?
Any suggestions will be most appreciated.
Thanks,
Tom

Since the Z3801A is being discussed, I thought I'd ask about an issue I'm having with my unit.  I use my Z3801 as my working lab standard for the usual pieces of RF test equipment.  In the past year or so I've had the unit drop out of lock and go into standby mode.  Resetting/cycling power would bring it back into lock for a while, but it generally got worse and now stays in hold over mode.  LH (thanks Mark Sims!!) reports that everything is operating normally except that the it has a PLL unlock.  The one highly suspicious item is that the DAC is at 99.996902% - full output. The unit shows it has over 94.5K hours run time, so I suspect that the OCXO has aged to the point that the EFC can't pull it into lock.  I disassembled the OCXO to see if it had a trimmer capacitor like the standard 10811 units.  After removing the outer case and foam insulation, I see that the outer heater is one of the thin printed circuit serpentine heaters on what I think is Kapton.  That would need to be peeled off of the case to either get to the hole for the trimmer (if there is one) or to remove the cover for the inner case. What I don't understand is the purpose of what looks like another coil or heater wrapped around the Kapton printed circuit heater stuck to the inner case.  This second coil/heater is 2 layers of 1/8" thick red foam wrapped completely around the inner case, with fine copper wires wrapped over each layer.  At least I think there are wires on each layer.  This whole second heater is taped down and I don't want to dig any further until I know a bit more about what I'm doing or find that there isn't a trimmer on the double oven 10811 oscillators. Any ideas on the purpose of the outer heater (or whatever it is)?  Does the double oven 10811 have a manual trimmer that I can adjust to bring the oscillator back to the center of the EFC range?  Assuming that I can manually adjust the OCXO back into adjustment range, will there be any issues with the Z3801 performance, things like phase noise, short term stability, etc? Any suggestions will be most appreciated. Thanks, Tom
BK
Bob kb8tq
Sat, Mar 3, 2018 7:06 PM

Hi

First off some basics about OCXO’s.

In a single oven design, you have a heater that warms up the entire crystal and the guts of the oscillator. It is on all the time and
it gets things up to a temperature that makes sense for a given crystal. It can be adjusted based on manufacturing data or by
trial and error to match the characteristics of that crystal.

In a double oven design, you have two ovens that are on all the time. One heats up the other one. They both work together to
achieve the end result. The gain of one adds to the gain of the other to give an improved result. On some double ovens, the
entire heat range of the inner oven is only 10’s of degrees ….

In a boosted oven, you have a second heater to get things going when it is very cold. This is an unusual approach and rarely
seen. Its normally easier to just design a bit more power into the main oven circuit. In a boosted design, the boost heat goes
away in normal operation at typical temperatures. In normal operation, the gain of the boost circuit does not count.

SO ….

The oscillator in the Z3801 is a boosted 10811. It is boosted to allow them to hit a spec of -40C on the unit. At the time it was
designed, there was talk about mounting these things in un-heated boxes outdoors. After they got a bit further into all the
details of the designs … that part went away. The spec still hung around long enough to apply to very early designs.

The net result is that you can pretty much destroy the outer heater stuff and the oscillator will work fine. There is no need
for it in a typical lab. There are some alarm triggers that need to be wired “ok” when you do so. The details are in the archives.

But …

Best guess if your unit is at max EFC = the “real” heater on the 10811 has quit working. To get at that, you will need to dig
into the guts of the unit. Given the massive EFC on the Z3801 version of the oscillator, it would take a crazy amount of aging
to hit limit.

Bob

On Mar 3, 2018, at 1:37 PM, Tom Curlee tcurlee@sbcglobal.net wrote:

Since the Z3801A is being discussed, I thought I'd ask about an issue I'm having with my unit.  I use my Z3801 as my working lab standard for the usual pieces of RF test equipment.  In the past year or so I've had the unit drop out of lock and go into standby mode.  Resetting/cycling power would bring it back into lock for a while, but it generally got worse and now stays in hold over mode.  LH (thanks Mark Sims!!) reports that everything is operating normally except that the it has a PLL unlock.  The one highly suspicious item is that the DAC is at 99.996902% - full output.
The unit shows it has over 94.5K hours run time, so I suspect that the OCXO has aged to the point that the EFC can't pull it into lock.  I disassembled the OCXO to see if it had a trimmer capacitor like the standard 10811 units.  After removing the outer case and foam insulation, I see that the outer heater is one of the thin printed circuit serpentine heaters on what I think is Kapton.  That would need to be peeled off of the case to either get to the hole for the trimmer (if there is one) or to remove the cover for the inner case.
What I don't understand is the purpose of what looks like another coil or heater wrapped around the Kapton printed circuit heater stuck to the inner case.  This second coil/heater is 2 layers of 1/8" thick red foam wrapped completely around the inner case, with fine copper wires wrapped over each layer.  At least I think there are wires on each layer.  This whole second heater is taped down and I don't want to dig any further until I know a bit more about what I'm doing or find that there isn't a trimmer on the double oven 10811 oscillators.
Any ideas on the purpose of the outer heater (or whatever it is)?  Does the double oven 10811 have a manual trimmer that I can adjust to bring the oscillator back to the center of the EFC range?  Assuming that I can manually adjust the OCXO back into adjustment range, will there be any issues with the Z3801 performance, things like phase noise, short term stability, etc?
Any suggestions will be most appreciated.
Thanks,
Tom


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi First off some basics about OCXO’s. In a single oven design, you have a heater that warms up the entire crystal and the guts of the oscillator. It is on all the time and it gets things up to a temperature that makes sense for a given crystal. It can be adjusted based on manufacturing data or by trial and error to match the characteristics of that crystal. In a double oven design, you have two ovens that are on all the time. One heats up the other one. They both work together to achieve the end result. The gain of one adds to the gain of the other to give an improved result. On some double ovens, the entire heat range of the inner oven is only 10’s of degrees …. In a boosted oven, you have a second heater to get things going when it is very cold. This is an unusual approach and rarely seen. Its normally easier to just design a bit more power into the main oven circuit. In a boosted design, the boost heat goes away in normal operation at typical temperatures. In normal operation, the gain of the boost circuit does not count. SO …. The oscillator in the Z3801 is a boosted 10811. It is boosted to allow them to hit a spec of -40C on the unit. At the time it was designed, there was talk about mounting these things in un-heated boxes outdoors. After they got a bit further into all the details of the designs … that part went away. The spec still hung around long enough to apply to very early designs. The net result is that you can pretty much destroy the outer heater stuff and the oscillator will work fine. There is no need for it in a typical lab. There are some alarm triggers that need to be wired “ok” when you do so. The details are in the archives. But … Best guess if your unit is at max EFC = the “real” heater on the 10811 has quit working. To get at that, you will need to dig into the guts of the unit. Given the massive EFC on the Z3801 version of the oscillator, it would take a crazy amount of aging to hit limit. Bob > On Mar 3, 2018, at 1:37 PM, Tom Curlee <tcurlee@sbcglobal.net> wrote: > > Since the Z3801A is being discussed, I thought I'd ask about an issue I'm having with my unit. I use my Z3801 as my working lab standard for the usual pieces of RF test equipment. In the past year or so I've had the unit drop out of lock and go into standby mode. Resetting/cycling power would bring it back into lock for a while, but it generally got worse and now stays in hold over mode. LH (thanks Mark Sims!!) reports that everything is operating normally except that the it has a PLL unlock. The one highly suspicious item is that the DAC is at 99.996902% - full output. > The unit shows it has over 94.5K hours run time, so I suspect that the OCXO has aged to the point that the EFC can't pull it into lock. I disassembled the OCXO to see if it had a trimmer capacitor like the standard 10811 units. After removing the outer case and foam insulation, I see that the outer heater is one of the thin printed circuit serpentine heaters on what I think is Kapton. That would need to be peeled off of the case to either get to the hole for the trimmer (if there is one) or to remove the cover for the inner case. > What I don't understand is the purpose of what looks like another coil or heater wrapped around the Kapton printed circuit heater stuck to the inner case. This second coil/heater is 2 layers of 1/8" thick red foam wrapped completely around the inner case, with fine copper wires wrapped over each layer. At least I think there are wires on each layer. This whole second heater is taped down and I don't want to dig any further until I know a bit more about what I'm doing or find that there isn't a trimmer on the double oven 10811 oscillators. > Any ideas on the purpose of the outer heater (or whatever it is)? Does the double oven 10811 have a manual trimmer that I can adjust to bring the oscillator back to the center of the EFC range? Assuming that I can manually adjust the OCXO back into adjustment range, will there be any issues with the Z3801 performance, things like phase noise, short term stability, etc? > Any suggestions will be most appreciated. > Thanks, > Tom > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
TH
Tom Holmes
Sun, Mar 4, 2018 3:12 AM

Bob...

Interesting point about the heater not working vs the XTAL having drifted too far. Mine has the same symptoms as the others reported (EFC at the end of its rope) but have not tackled it yet, figuring I'd have to dismantle the whole thing. Certainly troubleshooting a non-operating heater would be much more pleasant.

Thanks for that insight.

Tom Holmes, N8ZM

-----Original Message-----
From: time-nuts time-nuts-bounces@febo.com On Behalf Of Bob kb8tq
Sent: Saturday, March 03, 2018 2:07 PM
To: Tom Curlee tcurlee@sbcglobal.net; Discussion of precise time and frequency measurement time-nuts@febo.com
Subject: Re: [time-nuts] Z3801A OCXO manual trimming

Hi

First off some basics about OCXO’s.

In a single oven design, you have a heater that warms up the entire crystal and the guts of the oscillator. It is on all the time and
it gets things up to a temperature that makes sense for a given crystal. It can be adjusted based on manufacturing data or by
trial and error to match the characteristics of that crystal.

In a double oven design, you have two ovens that are on all the time. One heats up the other one. They both work together to
achieve the end result. The gain of one adds to the gain of the other to give an improved result. On some double ovens, the
entire heat range of the inner oven is only 10’s of degrees ….

In a boosted oven, you have a second heater to get things going when it is very cold. This is an unusual approach and rarely
seen. Its normally easier to just design a bit more power into the main oven circuit. In a boosted design, the boost heat goes
away in normal operation at typical temperatures. In normal operation, the gain of the boost circuit does not count.

SO ….

The oscillator in the Z3801 is a boosted 10811. It is boosted to allow them to hit a spec of -40C on the unit. At the time it was
designed, there was talk about mounting these things in un-heated boxes outdoors. After they got a bit further into all the
details of the designs … that part went away. The spec still hung around long enough to apply to very early designs.

The net result is that you can pretty much destroy the outer heater stuff and the oscillator will work fine. There is no need
for it in a typical lab. There are some alarm triggers that need to be wired “ok” when you do so. The details are in the archives.

But …

Best guess if your unit is at max EFC = the “real” heater on the 10811 has quit working. To get at that, you will need to dig
into the guts of the unit. Given the massive EFC on the Z3801 version of the oscillator, it would take a crazy amount of aging
to hit limit.

Bob

On Mar 3, 2018, at 1:37 PM, Tom Curlee tcurlee@sbcglobal.net wrote:

Since the Z3801A is being discussed, I thought I'd ask about an issue I'm having with my unit.  I use my Z3801 as my working lab standard for the usual pieces of RF test equipment.  In the past year or so I've had the unit drop out of lock and go into standby mode.  Resetting/cycling power would bring it back into lock for a while, but it generally got worse and now stays in hold over mode.  LH (thanks Mark Sims!!) reports that everything is operating normally except that the it has a PLL unlock.  The one highly suspicious item is that the DAC is at 99.996902% - full output.
The unit shows it has over 94.5K hours run time, so I suspect that the OCXO has aged to the point that the EFC can't pull it into lock.  I disassembled the OCXO to see if it had a trimmer capacitor like the standard 10811 units.  After removing the outer case and foam insulation, I see that the outer heater is one of the thin printed circuit serpentine heaters on what I think is Kapton.  That would need to be peeled off of the case to either get to the hole for the trimmer (if there is one) or to remove the cover for the inner case.
What I don't understand is the purpose of what looks like another coil or heater wrapped around the Kapton printed circuit heater stuck to the inner case.  This second coil/heater is 2 layers of 1/8" thick red foam wrapped completely around the inner case, with fine copper wires wrapped over each layer.  At least I think there are wires on each layer.  This whole second heater is taped down and I don't want to dig any further until I know a bit more about what I'm doing or find that there isn't a trimmer on the double oven 10811 oscillators.
Any ideas on the purpose of the outer heater (or whatever it is)?  Does the double oven 10811 have a manual trimmer that I can adjust to bring the oscillator back to the center of the EFC range?  Assuming that I can manually adjust the OCXO back into adjustment range, will there be any issues with the Z3801 performance, things like phase noise, short term stability, etc?
Any suggestions will be most appreciated.
Thanks,
Tom


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Bob... Interesting point about the heater not working vs the XTAL having drifted too far. Mine has the same symptoms as the others reported (EFC at the end of its rope) but have not tackled it yet, figuring I'd have to dismantle the whole thing. Certainly troubleshooting a non-operating heater would be much more pleasant. Thanks for that insight. Tom Holmes, N8ZM -----Original Message----- From: time-nuts <time-nuts-bounces@febo.com> On Behalf Of Bob kb8tq Sent: Saturday, March 03, 2018 2:07 PM To: Tom Curlee <tcurlee@sbcglobal.net>; Discussion of precise time and frequency measurement <time-nuts@febo.com> Subject: Re: [time-nuts] Z3801A OCXO manual trimming Hi First off some basics about OCXO’s. In a single oven design, you have a heater that warms up the entire crystal and the guts of the oscillator. It is on all the time and it gets things up to a temperature that makes sense for a given crystal. It can be adjusted based on manufacturing data or by trial and error to match the characteristics of that crystal. In a double oven design, you have two ovens that are on all the time. One heats up the other one. They both work together to achieve the end result. The gain of one adds to the gain of the other to give an improved result. On some double ovens, the entire heat range of the inner oven is only 10’s of degrees …. In a boosted oven, you have a second heater to get things going when it is very cold. This is an unusual approach and rarely seen. Its normally easier to just design a bit more power into the main oven circuit. In a boosted design, the boost heat goes away in normal operation at typical temperatures. In normal operation, the gain of the boost circuit does not count. SO …. The oscillator in the Z3801 is a boosted 10811. It is boosted to allow them to hit a spec of -40C on the unit. At the time it was designed, there was talk about mounting these things in un-heated boxes outdoors. After they got a bit further into all the details of the designs … that part went away. The spec still hung around long enough to apply to very early designs. The net result is that you can pretty much destroy the outer heater stuff and the oscillator will work fine. There is no need for it in a typical lab. There are some alarm triggers that need to be wired “ok” when you do so. The details are in the archives. But … Best guess if your unit is at max EFC = the “real” heater on the 10811 has quit working. To get at that, you will need to dig into the guts of the unit. Given the massive EFC on the Z3801 version of the oscillator, it would take a crazy amount of aging to hit limit. Bob > On Mar 3, 2018, at 1:37 PM, Tom Curlee <tcurlee@sbcglobal.net> wrote: > > Since the Z3801A is being discussed, I thought I'd ask about an issue I'm having with my unit. I use my Z3801 as my working lab standard for the usual pieces of RF test equipment. In the past year or so I've had the unit drop out of lock and go into standby mode. Resetting/cycling power would bring it back into lock for a while, but it generally got worse and now stays in hold over mode. LH (thanks Mark Sims!!) reports that everything is operating normally except that the it has a PLL unlock. The one highly suspicious item is that the DAC is at 99.996902% - full output. > The unit shows it has over 94.5K hours run time, so I suspect that the OCXO has aged to the point that the EFC can't pull it into lock. I disassembled the OCXO to see if it had a trimmer capacitor like the standard 10811 units. After removing the outer case and foam insulation, I see that the outer heater is one of the thin printed circuit serpentine heaters on what I think is Kapton. That would need to be peeled off of the case to either get to the hole for the trimmer (if there is one) or to remove the cover for the inner case. > What I don't understand is the purpose of what looks like another coil or heater wrapped around the Kapton printed circuit heater stuck to the inner case. This second coil/heater is 2 layers of 1/8" thick red foam wrapped completely around the inner case, with fine copper wires wrapped over each layer. At least I think there are wires on each layer. This whole second heater is taped down and I don't want to dig any further until I know a bit more about what I'm doing or find that there isn't a trimmer on the double oven 10811 oscillators. > Any ideas on the purpose of the outer heater (or whatever it is)? Does the double oven 10811 have a manual trimmer that I can adjust to bring the oscillator back to the center of the EFC range? Assuming that I can manually adjust the OCXO back into adjustment range, will there be any issues with the Z3801 performance, things like phase noise, short term stability, etc? > Any suggestions will be most appreciated. > Thanks, > Tom > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
BG
Bruce Griffiths
Sun, Mar 4, 2018 3:18 AM

Monitoring the output frequency whilst the oven "heats" up may give additional clues as will monitoring the heater current.

Bruce

 On 04 March 2018 at 16:12 Tom Holmes <tholmes@woh.rr.com> wrote:

 Bob...

 Interesting point about the heater not working vs the XTAL having drifted too far. Mine has the same symptoms as the others reported (EFC at the end of its rope) but have not tackled it yet, figuring I'd have to dismantle the whole thing. Certainly troubleshooting a non-operating heater would be much more pleasant.

 Thanks for that insight.

 Tom Holmes, N8ZM

 -----Original Message-----
 From: time-nuts <time-nuts-bounces@febo.com> On Behalf Of Bob kb8tq
 Sent: Saturday, March 03, 2018 2:07 PM
 To: Tom Curlee <tcurlee@sbcglobal.net>; Discussion of precise time and frequency measurement <time-nuts@febo.com>
 Subject: Re: [time-nuts] Z3801A OCXO manual trimming

 Hi

 First off some basics about OCXO’s.

 In a single oven design, you have a heater that warms up the entire crystal and the guts of the oscillator. It is on all the time and
 it gets things up to a temperature that makes sense for a given crystal. It can be adjusted based on manufacturing data or by
 trial and error to match the characteristics of that crystal.

 In a double oven design, you have two ovens that are on all the time. One heats up the other one. They both work together to
 achieve the end result. The gain of one adds to the gain of the other to give an improved result. On some double ovens, the
 entire heat range of the inner oven is only 10’s of degrees ….

 In a boosted oven, you have a second heater to get things going when it is very cold. This is an unusual approach and rarely
 seen. Its normally easier to just design a bit more power into the main oven circuit. In a boosted design, the boost heat goes
 away in normal operation at typical temperatures. In normal operation, the gain of the boost circuit does not count.

 SO ….

 The oscillator in the Z3801 is a boosted 10811. It is boosted to allow them to hit a spec of -40C on the unit. At the time it was
 designed, there was talk about mounting these things in un-heated boxes outdoors. After they got a bit further into all the
 details of the designs … that part went away. The spec still hung around long enough to apply to very early designs.

 The net result is that you can pretty much destroy the outer heater stuff and the oscillator will work fine. There is no need
 for it in a typical lab. There are some alarm triggers that need to be wired “ok” when you do so. The details are in the archives.

 But …

 Best guess if your unit is at max EFC = the “real” heater on the 10811 has quit working. To get at that, you will need to dig
 into the guts of the unit. Given the massive EFC on the Z3801 version of the oscillator, it would take a crazy amount of aging
 to hit limit.

 Bob
     On Mar 3, 2018, at 1:37 PM, Tom Curlee <tcurlee@sbcglobal.net> wrote:

     Since the Z3801A is being discussed, I thought I'd ask about an issue I'm having with my unit. I use my Z3801 as my working lab standard for the usual pieces of RF test equipment. In the past year or so I've had the unit drop out of lock and go into standby mode. Resetting/cycling power would bring it back into lock for a while, but it generally got worse and now stays in hold over mode. LH (thanks Mark Sims!!) reports that everything is operating normally except that the it has a PLL unlock. The one highly suspicious item is that the DAC is at 99.996902% - full output.
     The unit shows it has over 94.5K hours run time, so I suspect that the OCXO has aged to the point that the EFC can't pull it into lock. I disassembled the OCXO to see if it had a trimmer capacitor like the standard 10811 units. After removing the outer case and foam insulation, I see that the outer heater is one of the thin printed circuit serpentine heaters on what I think is Kapton. That would need to be peeled off of the case to either get to the hole for the trimmer (if there is one) or to remove the cover for the inner case.
     What I don't understand is the purpose of what looks like another coil or heater wrapped around the Kapton printed circuit heater stuck to the inner case. This second coil/heater is 2 layers of 1/8" thick red foam wrapped completely around the inner case, with fine copper wires wrapped over each layer. At least I think there are wires on each layer. This whole second heater is taped down and I don't want to dig any further until I know a bit more about what I'm doing or find that there isn't a trimmer on the double oven 10811 oscillators.
     Any ideas on the purpose of the outer heater (or whatever it is)? Does the double oven 10811 have a manual trimmer that I can adjust to bring the oscillator back to the center of the EFC range? Assuming that I can manually adjust the OCXO back into adjustment range, will there be any issues with the Z3801 performance, things like phase noise, short term stability, etc?
     Any suggestions will be most appreciated.
     Thanks,
     Tom

     _______________________________________________
     time-nuts mailing list -- time-nuts@febo.com
     To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
     and follow the instructions there.
 _______________________________________________
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

 _______________________________________________
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
Monitoring the output frequency whilst the oven "heats" up may give additional clues as will monitoring the heater current. Bruce > > On 04 March 2018 at 16:12 Tom Holmes <tholmes@woh.rr.com> wrote: > > Bob... > > Interesting point about the heater not working vs the XTAL having drifted too far. Mine has the same symptoms as the others reported (EFC at the end of its rope) but have not tackled it yet, figuring I'd have to dismantle the whole thing. Certainly troubleshooting a non-operating heater would be much more pleasant. > > Thanks for that insight. > > Tom Holmes, N8ZM > > -----Original Message----- > From: time-nuts <time-nuts-bounces@febo.com> On Behalf Of Bob kb8tq > Sent: Saturday, March 03, 2018 2:07 PM > To: Tom Curlee <tcurlee@sbcglobal.net>; Discussion of precise time and frequency measurement <time-nuts@febo.com> > Subject: Re: [time-nuts] Z3801A OCXO manual trimming > > Hi > > First off some basics about OCXO’s. > > In a single oven design, you have a heater that warms up the entire crystal and the guts of the oscillator. It is on all the time and > it gets things up to a temperature that makes sense for a given crystal. It can be adjusted based on manufacturing data or by > trial and error to match the characteristics of that crystal. > > In a double oven design, you have two ovens that are on all the time. One heats up the other one. They both work together to > achieve the end result. The gain of one adds to the gain of the other to give an improved result. On some double ovens, the > entire heat range of the inner oven is only 10’s of degrees …. > > In a boosted oven, you have a second heater to get things going when it is very cold. This is an unusual approach and rarely > seen. Its normally easier to just design a bit more power into the main oven circuit. In a boosted design, the boost heat goes > away in normal operation at typical temperatures. In normal operation, the gain of the boost circuit does not count. > > SO …. > > The oscillator in the Z3801 is a boosted 10811. It is boosted to allow them to hit a spec of -40C on the unit. At the time it was > designed, there was talk about mounting these things in un-heated boxes outdoors. After they got a bit further into all the > details of the designs … that part went away. The spec still hung around long enough to apply to very early designs. > > The net result is that you can pretty much destroy the outer heater stuff and the oscillator will work fine. There is no need > for it in a typical lab. There are some alarm triggers that need to be wired “ok” when you do so. The details are in the archives. > > But … > > Best guess if your unit is at max EFC = the “real” heater on the 10811 has quit working. To get at that, you will need to dig > into the guts of the unit. Given the massive EFC on the Z3801 version of the oscillator, it would take a crazy amount of aging > to hit limit. > > Bob > > > > > > On Mar 3, 2018, at 1:37 PM, Tom Curlee <tcurlee@sbcglobal.net> wrote: > > > > Since the Z3801A is being discussed, I thought I'd ask about an issue I'm having with my unit. I use my Z3801 as my working lab standard for the usual pieces of RF test equipment. In the past year or so I've had the unit drop out of lock and go into standby mode. Resetting/cycling power would bring it back into lock for a while, but it generally got worse and now stays in hold over mode. LH (thanks Mark Sims!!) reports that everything is operating normally except that the it has a PLL unlock. The one highly suspicious item is that the DAC is at 99.996902% - full output. > > The unit shows it has over 94.5K hours run time, so I suspect that the OCXO has aged to the point that the EFC can't pull it into lock. I disassembled the OCXO to see if it had a trimmer capacitor like the standard 10811 units. After removing the outer case and foam insulation, I see that the outer heater is one of the thin printed circuit serpentine heaters on what I think is Kapton. That would need to be peeled off of the case to either get to the hole for the trimmer (if there is one) or to remove the cover for the inner case. > > What I don't understand is the purpose of what looks like another coil or heater wrapped around the Kapton printed circuit heater stuck to the inner case. This second coil/heater is 2 layers of 1/8" thick red foam wrapped completely around the inner case, with fine copper wires wrapped over each layer. At least I think there are wires on each layer. This whole second heater is taped down and I don't want to dig any further until I know a bit more about what I'm doing or find that there isn't a trimmer on the double oven 10811 oscillators. > > Any ideas on the purpose of the outer heater (or whatever it is)? Does the double oven 10811 have a manual trimmer that I can adjust to bring the oscillator back to the center of the EFC range? Assuming that I can manually adjust the OCXO back into adjustment range, will there be any issues with the Z3801 performance, things like phase noise, short term stability, etc? > > Any suggestions will be most appreciated. > > Thanks, > > Tom > > > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > > > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
TV
Tom Van Baak
Sun, Mar 4, 2018 3:25 AM

Interesting point about the heater not working vs. the XTAL having drifted too far.

Any logs you made of EFC percent over the past few months or years will help verify the off-the-rail theory.

Another thing to try -- turn-off the Z3801A for a couple of hours to let it cool. Disconnect the GPS antenna. Then power it up and monitor the frequency, say every 10 s or a minute, until it stabilizes. The shape of this warm-up curve will give you a hint if the oven(s) are working. Repeat the test with the outer oven disconnected to test the inner oven by itself.

/tvb

> Interesting point about the heater not working vs. the XTAL having drifted too far. Any logs you made of EFC percent over the past few months or years will help verify the off-the-rail theory. Another thing to try -- turn-off the Z3801A for a couple of hours to let it cool. Disconnect the GPS antenna. Then power it up and monitor the frequency, say every 10 s or a minute, until it stabilizes. The shape of this warm-up curve will give you a hint if the oven(s) are working. Repeat the test with the outer oven disconnected to test the inner oven by itself. /tvb
BK
Bob kb8tq
Sun, Mar 4, 2018 2:51 PM

Hi

Heater failure is pretty easy to spot. You will be way off frequency. No need for a
GPSDO in your counter to see if it’s out by 0.0001 Hz. You should be > 20 Hz low
in frequency. Since there are multiple heaters, there are some odd combinations
in a 3801. Your inner heater can be dead, but the outer heater still works. The unit
will look like it’s warming up but doing so really slowly.

Based on seeing a lot of 10811’s and a lot of modern OCXO’s, a crystal that ages out
is pretty rare. The “something else” category is way more likely when diving into one
that is off frequency. That wasn’t always true so you can go back far enough in time
(like into the 60’s) and find units that have aged out. They also aged out in the first
ten years after they left the factory. Things got a lot better over the years in terms of
long term precision crystal aging ….

Bob

On Mar 3, 2018, at 10:12 PM, Tom Holmes tholmes@woh.rr.com wrote:

Bob...

Interesting point about the heater not working vs the XTAL having drifted too far. Mine has the same symptoms as the others reported (EFC at the end of its rope) but have not tackled it yet, figuring I'd have to dismantle the whole thing. Certainly troubleshooting a non-operating heater would be much more pleasant.

Thanks for that insight.

Tom Holmes, N8ZM

-----Original Message-----
From: time-nuts time-nuts-bounces@febo.com On Behalf Of Bob kb8tq
Sent: Saturday, March 03, 2018 2:07 PM
To: Tom Curlee tcurlee@sbcglobal.net; Discussion of precise time and frequency measurement time-nuts@febo.com
Subject: Re: [time-nuts] Z3801A OCXO manual trimming

Hi

First off some basics about OCXO’s.

In a single oven design, you have a heater that warms up the entire crystal and the guts of the oscillator. It is on all the time and
it gets things up to a temperature that makes sense for a given crystal. It can be adjusted based on manufacturing data or by
trial and error to match the characteristics of that crystal.

In a double oven design, you have two ovens that are on all the time. One heats up the other one. They both work together to
achieve the end result. The gain of one adds to the gain of the other to give an improved result. On some double ovens, the
entire heat range of the inner oven is only 10’s of degrees ….

In a boosted oven, you have a second heater to get things going when it is very cold. This is an unusual approach and rarely
seen. Its normally easier to just design a bit more power into the main oven circuit. In a boosted design, the boost heat goes
away in normal operation at typical temperatures. In normal operation, the gain of the boost circuit does not count.

SO ….

The oscillator in the Z3801 is a boosted 10811. It is boosted to allow them to hit a spec of -40C on the unit. At the time it was
designed, there was talk about mounting these things in un-heated boxes outdoors. After they got a bit further into all the
details of the designs … that part went away. The spec still hung around long enough to apply to very early designs.

The net result is that you can pretty much destroy the outer heater stuff and the oscillator will work fine. There is no need
for it in a typical lab. There are some alarm triggers that need to be wired “ok” when you do so. The details are in the archives.

But …

Best guess if your unit is at max EFC = the “real” heater on the 10811 has quit working. To get at that, you will need to dig
into the guts of the unit. Given the massive EFC on the Z3801 version of the oscillator, it would take a crazy amount of aging
to hit limit.

Bob

On Mar 3, 2018, at 1:37 PM, Tom Curlee tcurlee@sbcglobal.net wrote:

Since the Z3801A is being discussed, I thought I'd ask about an issue I'm having with my unit.  I use my Z3801 as my working lab standard for the usual pieces of RF test equipment.  In the past year or so I've had the unit drop out of lock and go into standby mode.  Resetting/cycling power would bring it back into lock for a while, but it generally got worse and now stays in hold over mode.  LH (thanks Mark Sims!!) reports that everything is operating normally except that the it has a PLL unlock.  The one highly suspicious item is that the DAC is at 99.996902% - full output.
The unit shows it has over 94.5K hours run time, so I suspect that the OCXO has aged to the point that the EFC can't pull it into lock.  I disassembled the OCXO to see if it had a trimmer capacitor like the standard 10811 units.  After removing the outer case and foam insulation, I see that the outer heater is one of the thin printed circuit serpentine heaters on what I think is Kapton.  That would need to be peeled off of the case to either get to the hole for the trimmer (if there is one) or to remove the cover for the inner case.
What I don't understand is the purpose of what looks like another coil or heater wrapped around the Kapton printed circuit heater stuck to the inner case.  This second coil/heater is 2 layers of 1/8" thick red foam wrapped completely around the inner case, with fine copper wires wrapped over each layer.  At least I think there are wires on each layer.  This whole second heater is taped down and I don't want to dig any further until I know a bit more about what I'm doing or find that there isn't a trimmer on the double oven 10811 oscillators.
Any ideas on the purpose of the outer heater (or whatever it is)?  Does the double oven 10811 have a manual trimmer that I can adjust to bring the oscillator back to the center of the EFC range?  Assuming that I can manually adjust the OCXO back into adjustment range, will there be any issues with the Z3801 performance, things like phase noise, short term stability, etc?
Any suggestions will be most appreciated.
Thanks,
Tom


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Heater failure is pretty easy to spot. You will be *way* off frequency. No need for a GPSDO in your counter to see if it’s out by 0.0001 Hz. You should be > 20 Hz low in frequency. Since there *are* multiple heaters, there are some odd combinations in a 3801. Your inner heater can be dead, but the outer heater still works. The unit will look like it’s warming up but doing so *really* slowly. Based on seeing a lot of 10811’s and a lot of modern OCXO’s, a crystal that ages out is pretty rare. The “something else” category is way more likely when diving into one that is off frequency. That wasn’t always true so you can go back far enough in time (like into the 60’s) and find units that have aged out. They also aged out in the first ten years after they left the factory. Things got a lot better over the years in terms of long term precision crystal aging …. Bob > On Mar 3, 2018, at 10:12 PM, Tom Holmes <tholmes@woh.rr.com> wrote: > > Bob... > > Interesting point about the heater not working vs the XTAL having drifted too far. Mine has the same symptoms as the others reported (EFC at the end of its rope) but have not tackled it yet, figuring I'd have to dismantle the whole thing. Certainly troubleshooting a non-operating heater would be much more pleasant. > > Thanks for that insight. > > Tom Holmes, N8ZM > > -----Original Message----- > From: time-nuts <time-nuts-bounces@febo.com> On Behalf Of Bob kb8tq > Sent: Saturday, March 03, 2018 2:07 PM > To: Tom Curlee <tcurlee@sbcglobal.net>; Discussion of precise time and frequency measurement <time-nuts@febo.com> > Subject: Re: [time-nuts] Z3801A OCXO manual trimming > > Hi > > First off some basics about OCXO’s. > > In a single oven design, you have a heater that warms up the entire crystal and the guts of the oscillator. It is on all the time and > it gets things up to a temperature that makes sense for a given crystal. It can be adjusted based on manufacturing data or by > trial and error to match the characteristics of that crystal. > > In a double oven design, you have two ovens that are on all the time. One heats up the other one. They both work together to > achieve the end result. The gain of one adds to the gain of the other to give an improved result. On some double ovens, the > entire heat range of the inner oven is only 10’s of degrees …. > > In a boosted oven, you have a second heater to get things going when it is very cold. This is an unusual approach and rarely > seen. Its normally easier to just design a bit more power into the main oven circuit. In a boosted design, the boost heat goes > away in normal operation at typical temperatures. In normal operation, the gain of the boost circuit does not count. > > SO …. > > The oscillator in the Z3801 is a boosted 10811. It is boosted to allow them to hit a spec of -40C on the unit. At the time it was > designed, there was talk about mounting these things in un-heated boxes outdoors. After they got a bit further into all the > details of the designs … that part went away. The spec still hung around long enough to apply to very early designs. > > The net result is that you can pretty much destroy the outer heater stuff and the oscillator will work fine. There is no need > for it in a typical lab. There are some alarm triggers that need to be wired “ok” when you do so. The details are in the archives. > > But … > > Best guess if your unit is at max EFC = the “real” heater on the 10811 has quit working. To get at that, you will need to dig > into the guts of the unit. Given the massive EFC on the Z3801 version of the oscillator, it would take a crazy amount of aging > to hit limit. > > Bob > > >> On Mar 3, 2018, at 1:37 PM, Tom Curlee <tcurlee@sbcglobal.net> wrote: >> >> Since the Z3801A is being discussed, I thought I'd ask about an issue I'm having with my unit. I use my Z3801 as my working lab standard for the usual pieces of RF test equipment. In the past year or so I've had the unit drop out of lock and go into standby mode. Resetting/cycling power would bring it back into lock for a while, but it generally got worse and now stays in hold over mode. LH (thanks Mark Sims!!) reports that everything is operating normally except that the it has a PLL unlock. The one highly suspicious item is that the DAC is at 99.996902% - full output. >> The unit shows it has over 94.5K hours run time, so I suspect that the OCXO has aged to the point that the EFC can't pull it into lock. I disassembled the OCXO to see if it had a trimmer capacitor like the standard 10811 units. After removing the outer case and foam insulation, I see that the outer heater is one of the thin printed circuit serpentine heaters on what I think is Kapton. That would need to be peeled off of the case to either get to the hole for the trimmer (if there is one) or to remove the cover for the inner case. >> What I don't understand is the purpose of what looks like another coil or heater wrapped around the Kapton printed circuit heater stuck to the inner case. This second coil/heater is 2 layers of 1/8" thick red foam wrapped completely around the inner case, with fine copper wires wrapped over each layer. At least I think there are wires on each layer. This whole second heater is taped down and I don't want to dig any further until I know a bit more about what I'm doing or find that there isn't a trimmer on the double oven 10811 oscillators. >> Any ideas on the purpose of the outer heater (or whatever it is)? Does the double oven 10811 have a manual trimmer that I can adjust to bring the oscillator back to the center of the EFC range? Assuming that I can manually adjust the OCXO back into adjustment range, will there be any issues with the Z3801 performance, things like phase noise, short term stability, etc? >> Any suggestions will be most appreciated. >> Thanks, >> Tom >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
R
rfnuts
Mon, Mar 5, 2018 11:51 PM

The easiest solution would be to drill a hole in the right place in
order to access the screw that covers the trimmer hole of the inner OCXO
housing.

Does anyone have a disassembled double oven 10811 at hands and could
take the required measures plus check if any vital parts of the outer
oven are in the way?

Adrian

Am 03.03.2018 um 19:37 schrieb Tom Curlee:

Since the Z3801A is being discussed, I thought I'd ask about an issue I'm having with my unit.  I use my Z3801 as my working lab standard for the usual pieces of RF test equipment.  In the past year or so I've had the unit drop out of lock and go into standby mode.  Resetting/cycling power would bring it back into lock for a while, but it generally got worse and now stays in hold over mode.  LH (thanks Mark Sims!!) reports that everything is operating normally except that the it has a PLL unlock.  The one highly suspicious item is that the DAC is at 99.996902% - full output.
The unit shows it has over 94.5K hours run time, so I suspect that the OCXO has aged to the point that the EFC can't pull it into lock.  I disassembled the OCXO to see if it had a trimmer capacitor like the standard 10811 units.  After removing the outer case and foam insulation, I see that the outer heater is one of the thin printed circuit serpentine heaters on what I think is Kapton.  That would need to be peeled off of the case to either get to the hole for the trimmer (if there is one) or to remove the cover for the inner case.
What I don't understand is the purpose of what looks like another coil or heater wrapped around the Kapton printed circuit heater stuck to the inner case.  This second coil/heater is 2 layers of 1/8" thick red foam wrapped completely around the inner case, with fine copper wires wrapped over each layer.  At least I think there are wires on each layer.  This whole second heater is taped down and I don't want to dig any further until I know a bit more about what I'm doing or find that there isn't a trimmer on the double oven 10811 oscillators.
Any ideas on the purpose of the outer heater (or whatever it is)?  Does the double oven 10811 have a manual trimmer that I can adjust to bring the oscillator back to the center of the EFC range?  Assuming that I can manually adjust the OCXO back into adjustment range, will there be any issues with the Z3801 performance, things like phase noise, short term stability, etc?
Any suggestions will be most appreciated.
Thanks,
Tom


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

The easiest solution would be to drill a hole in the right place in order to access the screw that covers the trimmer hole of the inner OCXO housing. Does anyone have a disassembled double oven 10811 at hands and could take the required measures plus check if any vital parts of the outer oven are in the way? Adrian Am 03.03.2018 um 19:37 schrieb Tom Curlee: > Since the Z3801A is being discussed, I thought I'd ask about an issue I'm having with my unit.  I use my Z3801 as my working lab standard for the usual pieces of RF test equipment.  In the past year or so I've had the unit drop out of lock and go into standby mode.  Resetting/cycling power would bring it back into lock for a while, but it generally got worse and now stays in hold over mode.  LH (thanks Mark Sims!!) reports that everything is operating normally except that the it has a PLL unlock.  The one highly suspicious item is that the DAC is at 99.996902% - full output. > The unit shows it has over 94.5K hours run time, so I suspect that the OCXO has aged to the point that the EFC can't pull it into lock.  I disassembled the OCXO to see if it had a trimmer capacitor like the standard 10811 units.  After removing the outer case and foam insulation, I see that the outer heater is one of the thin printed circuit serpentine heaters on what I think is Kapton.  That would need to be peeled off of the case to either get to the hole for the trimmer (if there is one) or to remove the cover for the inner case. > What I don't understand is the purpose of what looks like another coil or heater wrapped around the Kapton printed circuit heater stuck to the inner case.  This second coil/heater is 2 layers of 1/8" thick red foam wrapped completely around the inner case, with fine copper wires wrapped over each layer.  At least I think there are wires on each layer.  This whole second heater is taped down and I don't want to dig any further until I know a bit more about what I'm doing or find that there isn't a trimmer on the double oven 10811 oscillators. > Any ideas on the purpose of the outer heater (or whatever it is)?  Does the double oven 10811 have a manual trimmer that I can adjust to bring the oscillator back to the center of the EFC range?  Assuming that I can manually adjust the OCXO back into adjustment range, will there be any issues with the Z3801 performance, things like phase noise, short term stability, etc? > Any suggestions will be most appreciated. > Thanks, > Tom > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >