Would you be concerned if the manufacturer does not have an uncertainty budget, so can't provide uncertainties in a calibration?

DD
Dr. David Kirkby
Thu, Apr 19, 2018 12:08 PM

I have an HP 4339B high resistance meter. It can read up to 1.6 x 10^16
ohms, with a basic uncertainty of 0.6%. It has a built in voltage source of
up to 1 kV.

I've contacted Keysight (UK) and asked for calibration cost, with
uncertainties, for this 4339B. However, they have said they can't provide a
calibration with uncertainties, and when I asked why, they have said they
do not have an uncertainty budget available that suites that model. Looking
at the Keysight website, a calibration with uncertainties is available in
the USA, but I guess for whatever reason Keysight UK don't have this
ability on this specific instrument. On other instruments I have sent them,
I have never had this issue.

I expect if I really wanted to, I could get it shipped to the USA and
calibrated there, but I can't justify the costs that would be incurred if
it was shipped to the USA and back.

From a practical perspective, I don't really need the uncertainties - it

was more for interest sake. I also have a reasonable degree of confidence
that as a reputable company, Keysight would not calibrate an instrument
unless they were confident they could determine if it is in or out of
specification.

The 4339B is a pretty obscure unit, requiring resistors up to 10^11 ohms to
calibrate it.

I'm sending this to Keysight with a blank EEPROM, so there will be no
calibration data whatsoever in the instrument. Hopefully that means
everything will be set right, and so likely to stay in specification longer
than it might otherwise do so. There are no trimmers in the 4339B - all
calibration is via the EEPROM. A 3458A is used for calibration of the
voltage source. I'm confidence the voltages will be measured accurately
enough, but a bit less confident about the values of the resistors used for
calibration.

Dave

Dr David Kirkby Ph.D C.Eng MIET
Kirkby Microwave Ltd
Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, CHELMSFORD,
Essex, CM3 6DT, United Kingdom.
Registered in England and Wales as company number 08914892
http://www.kirkbymicrowave.co.uk/
Tel 01621-680100 / +44 1621-680100

I have an HP 4339B high resistance meter. It can read up to 1.6 x 10^16 ohms, with a basic uncertainty of 0.6%. It has a built in voltage source of up to 1 kV. I've contacted Keysight (UK) and asked for calibration cost, with uncertainties, for this 4339B. However, they have said they can't provide a calibration with uncertainties, and when I asked why, they have said they do not have an uncertainty budget available that suites that model. Looking at the Keysight website, a calibration with uncertainties is available in the USA, but I guess for whatever reason Keysight UK don't have this ability on this specific instrument. On other instruments I have sent them, I have never had this issue. I expect if I really wanted to, I could get it shipped to the USA and calibrated there, but I can't justify the costs that would be incurred if it was shipped to the USA and back. >From a practical perspective, I don't really need the uncertainties - it was more for interest sake. I also have a reasonable degree of confidence that as a reputable company, Keysight would not calibrate an instrument unless they were confident they could determine if it is in or out of specification. The 4339B is a pretty obscure unit, requiring resistors up to 10^11 ohms to calibrate it. I'm sending this to Keysight with a blank EEPROM, so there will be no calibration data whatsoever in the instrument. Hopefully that means everything will be set right, and so likely to stay in specification longer than it might otherwise do so. There are no trimmers in the 4339B - all calibration is via the EEPROM. A 3458A is used for calibration of the voltage source. I'm confidence the voltages will be measured accurately enough, but a bit less confident about the values of the resistors used for calibration. Dave Dr David Kirkby Ph.D C.Eng MIET Kirkby Microwave Ltd Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, CHELMSFORD, Essex, CM3 6DT, United Kingdom. Registered in England and Wales as company number 08914892 http://www.kirkbymicrowave.co.uk/ Tel 01621-680100 / +44 1621-680100
A
acbern@gmx.de
Thu, Apr 19, 2018 3:42 PM

-a calibration certificate without uncertainsties is totally useless. in is not even a calibration.
-I have never understood why people are so keen on getting things calibrated at Keysight.

Gesendet: Donnerstag, 19. April 2018 um 14:08 Uhr
Von: "Dr. David Kirkby" drkirkby@kirkbymicrowave.co.uk
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: [volt-nuts] Would you be concerned if the manufacturer does not have an uncertainty budget, so can't provide uncertainties in a calibration?

I have an HP 4339B high resistance meter. It can read up to 1.6 x 10^16
ohms, with a basic uncertainty of 0.6%. It has a built in voltage source of
up to 1 kV.

I've contacted Keysight (UK) and asked for calibration cost, with
uncertainties, for this 4339B. However, they have said they can't provide a
calibration with uncertainties, and when I asked why, they have said they
do not have an uncertainty budget available that suites that model. Looking
at the Keysight website, a calibration with uncertainties is available in
the USA, but I guess for whatever reason Keysight UK don't have this
ability on this specific instrument. On other instruments I have sent them,
I have never had this issue.

I expect if I really wanted to, I could get it shipped to the USA and
calibrated there, but I can't justify the costs that would be incurred if
it was shipped to the USA and back.

From a practical perspective, I don't really need the uncertainties - it
was more for interest sake. I also have a reasonable degree of confidence
that as a reputable company, Keysight would not calibrate an instrument
unless they were confident they could determine if it is in or out of
specification.

The 4339B is a pretty obscure unit, requiring resistors up to 10^11 ohms to
calibrate it.

I'm sending this to Keysight with a blank EEPROM, so there will be no
calibration data whatsoever in the instrument. Hopefully that means
everything will be set right, and so likely to stay in specification longer
than it might otherwise do so. There are no trimmers in the 4339B - all
calibration is via the EEPROM. A 3458A is used for calibration of the
voltage source. I'm confidence the voltages will be measured accurately
enough, but a bit less confident about the values of the resistors used for
calibration.

Dave

Dr David Kirkby Ph.D C.Eng MIET
Kirkby Microwave Ltd
Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, CHELMSFORD,
Essex, CM3 6DT, United Kingdom.
Registered in England and Wales as company number 08914892
http://www.kirkbymicrowave.co.uk/
Tel 01621-680100 / +44 1621-680100


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

-a calibration certificate without uncertainsties is totally useless. in is not even a calibration. -I have never understood why people are so keen on getting things calibrated at Keysight. > Gesendet: Donnerstag, 19. April 2018 um 14:08 Uhr > Von: "Dr. David Kirkby" <drkirkby@kirkbymicrowave.co.uk> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > Betreff: [volt-nuts] Would you be concerned if the manufacturer does not have an uncertainty budget, so can't provide uncertainties in a calibration? > > I have an HP 4339B high resistance meter. It can read up to 1.6 x 10^16 > ohms, with a basic uncertainty of 0.6%. It has a built in voltage source of > up to 1 kV. > > I've contacted Keysight (UK) and asked for calibration cost, with > uncertainties, for this 4339B. However, they have said they can't provide a > calibration with uncertainties, and when I asked why, they have said they > do not have an uncertainty budget available that suites that model. Looking > at the Keysight website, a calibration with uncertainties is available in > the USA, but I guess for whatever reason Keysight UK don't have this > ability on this specific instrument. On other instruments I have sent them, > I have never had this issue. > > I expect if I really wanted to, I could get it shipped to the USA and > calibrated there, but I can't justify the costs that would be incurred if > it was shipped to the USA and back. > > From a practical perspective, I don't really need the uncertainties - it > was more for interest sake. I also have a reasonable degree of confidence > that as a reputable company, Keysight would not calibrate an instrument > unless they were confident they could determine if it is in or out of > specification. > > The 4339B is a pretty obscure unit, requiring resistors up to 10^11 ohms to > calibrate it. > > I'm sending this to Keysight with a blank EEPROM, so there will be no > calibration data whatsoever in the instrument. Hopefully that means > everything will be set right, and so likely to stay in specification longer > than it might otherwise do so. There are no trimmers in the 4339B - all > calibration is via the EEPROM. A 3458A is used for calibration of the > voltage source. I'm confidence the voltages will be measured accurately > enough, but a bit less confident about the values of the resistors used for > calibration. > > Dave > > Dr David Kirkby Ph.D C.Eng MIET > Kirkby Microwave Ltd > Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, CHELMSFORD, > Essex, CM3 6DT, United Kingdom. > Registered in England and Wales as company number 08914892 > http://www.kirkbymicrowave.co.uk/ > Tel 01621-680100 / +44 1621-680100 > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >
DD
Dr. David Kirkby
Thu, Apr 19, 2018 4:41 PM

On 19 April 2018 at 16:42, acbern@gmx.de wrote:

-a calibration certificate without uncertainsties is totally useless. in
is not even a calibration.

Having bought the meter, it is not so useless if it tells me it is working
or not, but I do have some concerns I must admit.

-I have never understood why people are so keen on getting things
calibrated at Keysight.

Well, quite simply there's nobody else I would trust to calibrate much of
the Agilent equipment. I did contact one UKAS acredited lab, who quoted to
calibrate loads of bits of my equipment, but declined this meter. But when
I checked the companies uncertainties, I was totally unimpressed. For
example, their uncertainty on capacitance at 1 MHz was well in excess of
0.05%, yet they quoted to calibrate the meter, which has a basic
uncertainty of 0.05%. I also found their prices were much higher than
Keysight.

Most companies are not going to be able to adjust Agilent stuff if it is
out of spec anyway, as often the software to make the adjustments is not
available. So I'm not convinced there is any half-sensible alternative.

Dave

On 19 April 2018 at 16:42, <acbern@gmx.de> wrote: > -a calibration certificate without uncertainsties is totally useless. in > is not even a calibration. > Having bought the meter, it is not so useless if it tells me it is working or not, but I do have some concerns I must admit. > -I have never understood why people are so keen on getting things > calibrated at Keysight. > Well, quite simply there's nobody else I would trust to calibrate much of the Agilent equipment. I did contact one UKAS acredited lab, who quoted to calibrate loads of bits of my equipment, but declined this meter. But when I checked the companies uncertainties, I was totally unimpressed. For example, their uncertainty on capacitance at 1 MHz was well in excess of 0.05%, yet they quoted to calibrate the meter, which has a basic uncertainty of 0.05%. I also found their prices were much higher than Keysight. Most companies are not going to be able to adjust Agilent stuff if it is out of spec anyway, as often the software to make the adjustments is not available. So I'm not convinced there is any half-sensible alternative. Dave
A
acbern@gmx.de
Fri, Apr 20, 2018 5:53 AM

understand, it is maybe different here in Germany then, there are a number of labs with pretty low uncertainties who also support old gear adjustments.
on the other hand, you have to know who you are working with for what gear. Otherwise you can have unpleasant surprises.
in your case, if you have no nobody calibrating your meter with a specified uncertainty, maybe an option is to build your own decade of resistances and use a precisiion high voltage source (calibrator) and a 3458A as current meter. that brings you to say 10nA/100Gohms. if thats sufficient.

Gesendet: Donnerstag, 19. April 2018 um 18:41 Uhr
Von: "Dr. David Kirkby" drkirkby@kirkbymicrowave.co.uk
An: "Discussion of precise voltage measurement" volt-nuts@febo.com
Betreff: Re: [volt-nuts] Would you be concerned if the manufacturer does not have an uncertainty budget, so can't provide uncertainties in a calibration?

On 19 April 2018 at 16:42, acbern@gmx.de wrote:

-a calibration certificate without uncertainsties is totally useless. in
is not even a calibration.

Having bought the meter, it is not so useless if it tells me it is working
or not, but I do have some concerns I must admit.

-I have never understood why people are so keen on getting things
calibrated at Keysight.

Well, quite simply there's nobody else I would trust to calibrate much of
the Agilent equipment. I did contact one UKAS acredited lab, who quoted to
calibrate loads of bits of my equipment, but declined this meter. But when
I checked the companies uncertainties, I was totally unimpressed. For
example, their uncertainty on capacitance at 1 MHz was well in excess of
0.05%, yet they quoted to calibrate the meter, which has a basic
uncertainty of 0.05%. I also found their prices were much higher than
Keysight.

Most companies are not going to be able to adjust Agilent stuff if it is
out of spec anyway, as often the software to make the adjustments is not
available. So I'm not convinced there is any half-sensible alternative.

Dave


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

understand, it is maybe different here in Germany then, there are a number of labs with pretty low uncertainties who also support old gear adjustments. on the other hand, you have to know who you are working with for what gear. Otherwise you can have unpleasant surprises. in your case, if you have no nobody calibrating your meter with a specified uncertainty, maybe an option is to build your own decade of resistances and use a precisiion high voltage source (calibrator) and a 3458A as current meter. that brings you to say 10nA/100Gohms. if thats sufficient. > Gesendet: Donnerstag, 19. April 2018 um 18:41 Uhr > Von: "Dr. David Kirkby" <drkirkby@kirkbymicrowave.co.uk> > An: "Discussion of precise voltage measurement" <volt-nuts@febo.com> > Betreff: Re: [volt-nuts] Would you be concerned if the manufacturer does not have an uncertainty budget, so can't provide uncertainties in a calibration? > > On 19 April 2018 at 16:42, <acbern@gmx.de> wrote: > > > -a calibration certificate without uncertainsties is totally useless. in > > is not even a calibration. > > > > Having bought the meter, it is not so useless if it tells me it is working > or not, but I do have some concerns I must admit. > > > > -I have never understood why people are so keen on getting things > > calibrated at Keysight. > > > > Well, quite simply there's nobody else I would trust to calibrate much of > the Agilent equipment. I did contact one UKAS acredited lab, who quoted to > calibrate loads of bits of my equipment, but declined this meter. But when > I checked the companies uncertainties, I was totally unimpressed. For > example, their uncertainty on capacitance at 1 MHz was well in excess of > 0.05%, yet they quoted to calibrate the meter, which has a basic > uncertainty of 0.05%. I also found their prices were much higher than > Keysight. > > Most companies are not going to be able to adjust Agilent stuff if it is > out of spec anyway, as often the software to make the adjustments is not > available. So I'm not convinced there is any half-sensible alternative. > > Dave > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >
DD
Dr. David Kirkby
Fri, Apr 20, 2018 1:37 PM

On 19 April 2018 at 16:42, acbern@gmx.de wrote:

-a calibration certificate without uncertainsties is totally useless. in
is not even a calibration.
-I have never understood why people are so keen on getting things
calibrated at Keysight.

I must admit I do have some concerns, about this. There are 3 aspects to
the calibration

  1. Calibrate the voltage of the internal source using a 3458A. I have no
    concerns a 3458A is not more than capable of measuring the voltage.

  2. Measure the resistors in the 16340A. What puzzles me here, is that when
    I had another 4339B calibrated, the limits on measuring those resistors
    were symmetrical about the nominal values of 1e6, 1e7, 1e8, 1e8, 1e10 and
    1e11 ohms. I would have expected the limits to be asymmetrical, because
    those resistors are probably not their nominal value.

The columns below, from left to right are

Device type (whether the DUT is floating, or grounded one side).
Resistor setting (ohms)
Votage (V)
Measurement time (Long or Short)
Test limits (+/- ohm)
Test results (ohm)

FLOAT 1E6 100 SHORT +/- 0.0086E6 -.0019E6
FLOAT 1E7 100 LONG +/- 0.0063E7 -.0016E7
FLOAT 1E8 100 LONG +/- 0.0073E8 -.0027E8
FLOAT 1E9 100 LONG +/- 0.0093E9 -.0032E9
FLOAT 1E10 100 LONG +/- 0.0273E10 +.0095E10
FLOAT 1E11 100 LONG +/- 0.0453E11 +.0080E11
FLOAT 1E11 100 SHORT +/- 0.0550E11 +.0086E11
FLOAT 1E11 10 LONG +/- 0.0546E11 +.0113E11
GROUND 1E7 100 LONG +/- 0.0065E7 -.0017E7
GROUND 1E11 10 LONG +/- 0.0573E11 +.0107E11

That strikes me that the assumption is the values are what their nominal
values are, but I wonder how accurate they are.

  1. Apply known currents, again using the test box.

I've asked Keysight if they are certain that the meter can be put within
specification, and if not whether it might be sent outside the UK to be
calibrated.

To be honest, I don't really NEED the uncertainties, but the fact they
can't provide them does concern me a bit.

Dr David Kirkby Ph.D C.Eng MIET
Kirkby Microwave Ltd
Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, CHELMSFORD,
Essex, CM3 6DT, United Kingdom.
Registered in England and Wales as company number 08914892
http://www.kirkbymicrowave.co.uk/
Tel 01621-680100 / +44 1621-680100

On 19 April 2018 at 16:42, <acbern@gmx.de> wrote: > -a calibration certificate without uncertainsties is totally useless. in > is not even a calibration. > -I have never understood why people are so keen on getting things > calibrated at Keysight. > I must admit I do have some concerns, about this. There are 3 aspects to the calibration 1) Calibrate the voltage of the internal source using a 3458A. I have no concerns a 3458A is not more than capable of measuring the voltage. 2) Measure the resistors in the 16340A. What puzzles me here, is that when I had another 4339B calibrated, the limits on measuring those resistors were symmetrical about the nominal values of 1e6, 1e7, 1e8, 1e8, 1e10 and 1e11 ohms. I would have expected the limits to be asymmetrical, because those resistors are probably not their nominal value. The columns below, from left to right are Device type (whether the DUT is floating, or grounded one side). Resistor setting (ohms) Votage (V) Measurement time (Long or Short) Test limits (+/- ohm) Test results (ohm) FLOAT 1E6 100 SHORT +/- 0.0086E6 -.0019E6 FLOAT 1E7 100 LONG +/- 0.0063E7 -.0016E7 FLOAT 1E8 100 LONG +/- 0.0073E8 -.0027E8 FLOAT 1E9 100 LONG +/- 0.0093E9 -.0032E9 FLOAT 1E10 100 LONG +/- 0.0273E10 +.0095E10 FLOAT 1E11 100 LONG +/- 0.0453E11 +.0080E11 FLOAT 1E11 100 SHORT +/- 0.0550E11 +.0086E11 FLOAT 1E11 10 LONG +/- 0.0546E11 +.0113E11 GROUND 1E7 100 LONG +/- 0.0065E7 -.0017E7 GROUND 1E11 10 LONG +/- 0.0573E11 +.0107E11 That strikes me that the assumption is the values are what their nominal values are, but I wonder how accurate they are. 3) Apply known currents, again using the test box. I've asked Keysight if they are certain that the meter can be put within specification, and if not whether it might be sent outside the UK to be calibrated. To be honest, I don't really NEED the uncertainties, but the fact they can't provide them does concern me a bit. Dr David Kirkby Ph.D C.Eng MIET Kirkby Microwave Ltd Registered office: Stokes Hall Lodge, Burnham Rd, Althorne, CHELMSFORD, Essex, CM3 6DT, United Kingdom. Registered in England and Wales as company number 08914892 http://www.kirkbymicrowave.co.uk/ Tel 01621-680100 / +44 1621-680100
FT
Florian Teply
Sat, Apr 21, 2018 8:32 AM

Am Fri, 20 Apr 2018 14:37:22 +0100
schrieb "Dr. David Kirkby" drkirkby@kirkbymicrowave.co.uk:

  1. Measure the resistors in the 16340A. What puzzles me here, is that
    when I had another 4339B calibrated, the limits on measuring those
    resistors were symmetrical about the nominal values of 1e6, 1e7, 1e8,
    1e8, 1e10 and 1e11 ohms. I would have expected the limits to be
    asymmetrical, because those resistors are probably not their nominal
    value.

The columns below, from left to right are

Device type (whether the DUT is floating, or grounded one side).
Resistor setting (ohms)
Votage (V)
Measurement time (Long or Short)
Test limits (+/- ohm)
Test results (ohm)

FLOAT 1E6 100 SHORT +/- 0.0086E6 -.0019E6
FLOAT 1E7 100 LONG +/- 0.0063E7 -.0016E7
FLOAT 1E8 100 LONG +/- 0.0073E8 -.0027E8
FLOAT 1E9 100 LONG +/- 0.0093E9 -.0032E9
FLOAT 1E10 100 LONG +/- 0.0273E10 +.0095E10
FLOAT 1E11 100 LONG +/- 0.0453E11 +.0080E11
FLOAT 1E11 100 SHORT +/- 0.0550E11 +.0086E11
FLOAT 1E11 10 LONG +/- 0.0546E11 +.0113E11
GROUND 1E7 100 LONG +/- 0.0065E7 -.0017E7
GROUND 1E11 10 LONG +/- 0.0573E11 +.0107E11

That strikes me that the assumption is the values are what their
nominal values are, but I wonder how accurate they are.

I might be wrong, but to me it seems like the resistors are not
exactly nominal but slightly off. But the uncertainty of the
measurements is larger than the deviation. Or were you referring to the
notion that the uncertainties are symmetrically distributed? I'd be
pretty surprised if the uncertainties were asymmetrical for that
matter.

Best regards,
Florian

Am Fri, 20 Apr 2018 14:37:22 +0100 schrieb "Dr. David Kirkby" <drkirkby@kirkbymicrowave.co.uk>: > 2) Measure the resistors in the 16340A. What puzzles me here, is that > when I had another 4339B calibrated, the limits on measuring those > resistors were symmetrical about the nominal values of 1e6, 1e7, 1e8, > 1e8, 1e10 and 1e11 ohms. I would have expected the limits to be > asymmetrical, because those resistors are probably not their nominal > value. > > The columns below, from left to right are > > Device type (whether the DUT is floating, or grounded one side). > Resistor setting (ohms) > Votage (V) > Measurement time (Long or Short) > Test limits (+/- ohm) > Test results (ohm) > > FLOAT 1E6 100 SHORT +/- 0.0086E6 -.0019E6 > FLOAT 1E7 100 LONG +/- 0.0063E7 -.0016E7 > FLOAT 1E8 100 LONG +/- 0.0073E8 -.0027E8 > FLOAT 1E9 100 LONG +/- 0.0093E9 -.0032E9 > FLOAT 1E10 100 LONG +/- 0.0273E10 +.0095E10 > FLOAT 1E11 100 LONG +/- 0.0453E11 +.0080E11 > FLOAT 1E11 100 SHORT +/- 0.0550E11 +.0086E11 > FLOAT 1E11 10 LONG +/- 0.0546E11 +.0113E11 > GROUND 1E7 100 LONG +/- 0.0065E7 -.0017E7 > GROUND 1E11 10 LONG +/- 0.0573E11 +.0107E11 > > That strikes me that the assumption is the values are what their > nominal values are, but I wonder how accurate they are. > I might be wrong, but to me it seems like the resistors are not exactly nominal but slightly off. But the uncertainty of the measurements is larger than the deviation. Or were you referring to the notion that the uncertainties are symmetrically distributed? I'd be pretty surprised if the uncertainties were asymmetrical for that matter. Best regards, Florian
DD
Dr. David Kirkby
Mon, Apr 23, 2018 10:59 PM

On 21 April 2018 at 09:32, Florian Teply usenet@teply.info wrote:

Am Fri, 20 Apr 2018 14:37:22 +0100
schrieb "Dr. David Kirkby" drkirkby@kirkbymicrowave.co.uk:

The columns below, from left to right are

Device type (whether the DUT is floating, or grounded one side).
Resistor setting (ohms)
Votage (V)
Measurement time (Long or Short)
Test limits (+/- ohm)
Test results (ohm)

FLOAT 1E6 100 SHORT +/- 0.0086E6 -.0019E6
FLOAT 1E7 100 LONG +/- 0.0063E7 -.0016E7
FLOAT 1E8 100 LONG +/- 0.0073E8 -.0027E8
FLOAT 1E9 100 LONG +/- 0.0093E9 -.0032E9
FLOAT 1E10 100 LONG +/- 0.0273E10 +.0095E10
FLOAT 1E11 100 LONG +/- 0.0453E11 +.0080E11
FLOAT 1E11 100 SHORT +/- 0.0550E11 +.0086E11
FLOAT 1E11 10 LONG +/- 0.0546E11 +.0113E11
GROUND 1E7 100 LONG +/- 0.0065E7 -.0017E7
GROUND 1E11 10 LONG +/- 0.0573E11 +.0107E11

That strikes me that the assumption is the values are what their
nominal values are, but I wonder how accurate they are.

I might be wrong, but to me it seems like the resistors are not
exactly nominal but slightly off. But the uncertainty of the
measurements is larger than the deviation. Or were you referring to the
notion that the uncertainties are symmetrically distributed? I'd be
pretty surprised if the uncertainties were asymmetrical for that
matter.

Best regards,
Florian

Florian,

how do you determine that the resistors are not assumed to be the nominal
value?

As far as I can see, taking the example of a 1e11 ohm resistor grounded at
one end (very last entry on table), the meter should read 1e11 +/-
0.0573e11 ohms. My meter read 0.010e11 ohms high, so was in spec, as
0.010e11 is less than 0.0573e11. As far as I can determine, the fact the
permissable range of the meter is +/-x, rather than +x, -y, means the
nominal values are assumed.

I put the complete cal certificate here.

http://www.kirkbymicrowave.co.uk/tmp/1-9690444179-1-combined-file.pdf

What I also find a bit odd, is the 16340A RC box used for calibrating the
meter, is itself not due for calibration for over a year.

I am awaiting a call/email from the calibration manager at Keysight (UK),
and I've been advised Keysight (UK) have contacted Keysight in the USA to
see what they can provide, as a calibration with uncertainties is listed
on the Keysight (USA) website.

I have no formal requirement for needing the uncertainties, but I am a bit
worried the fact that Keysight (UK) seem to use a resistance box that is
calibrated less than once/year, and can't provide the uncertainties, and as
far as I can tell (although you disagree), it would appear the nominal
value of the resistors are used. . It does not exactly inspire a lot of
confidence.

Dave

On 21 April 2018 at 09:32, Florian Teply <usenet@teply.info> wrote: > Am Fri, 20 Apr 2018 14:37:22 +0100 > schrieb "Dr. David Kirkby" <drkirkby@kirkbymicrowave.co.uk>: > > > The columns below, from left to right are > > > > Device type (whether the DUT is floating, or grounded one side). > > Resistor setting (ohms) > > Votage (V) > > Measurement time (Long or Short) > > Test limits (+/- ohm) > > Test results (ohm) > > > > FLOAT 1E6 100 SHORT +/- 0.0086E6 -.0019E6 > > FLOAT 1E7 100 LONG +/- 0.0063E7 -.0016E7 > > FLOAT 1E8 100 LONG +/- 0.0073E8 -.0027E8 > > FLOAT 1E9 100 LONG +/- 0.0093E9 -.0032E9 > > FLOAT 1E10 100 LONG +/- 0.0273E10 +.0095E10 > > FLOAT 1E11 100 LONG +/- 0.0453E11 +.0080E11 > > FLOAT 1E11 100 SHORT +/- 0.0550E11 +.0086E11 > > FLOAT 1E11 10 LONG +/- 0.0546E11 +.0113E11 > > GROUND 1E7 100 LONG +/- 0.0065E7 -.0017E7 > > GROUND 1E11 10 LONG +/- 0.0573E11 +.0107E11 > > > > That strikes me that the assumption is the values are what their > > nominal values are, but I wonder how accurate they are. > > > I might be wrong, but to me it seems like the resistors are not > exactly nominal but slightly off. But the uncertainty of the > measurements is larger than the deviation. Or were you referring to the > notion that the uncertainties are symmetrically distributed? I'd be > pretty surprised if the uncertainties were asymmetrical for that > matter. > > Best regards, > Florian > Florian, how do you determine that the resistors are not assumed to be the nominal value? As far as I can see, taking the example of a 1e11 ohm resistor grounded at one end (very last entry on table), the meter should read 1e11 +/- 0.0573e11 ohms. My meter read 0.010e11 ohms high, so was in spec, as 0.010e11 is less than 0.0573e11. As far as I can determine, the fact the permissable range of the meter is +/-x, rather than +x, -y, means the nominal values are assumed. I put the complete cal certificate here. http://www.kirkbymicrowave.co.uk/tmp/1-9690444179-1-combined-file.pdf What I also find a bit odd, is the 16340A RC box used for calibrating the meter, is itself not due for calibration for over a year. I am awaiting a call/email from the calibration manager at Keysight (UK), and I've been advised Keysight (UK) have contacted Keysight in the USA to see what they can provide, as a calibration *with* uncertainties is listed on the Keysight (USA) website. I have no formal requirement for needing the uncertainties, but I am a bit worried the fact that Keysight (UK) seem to use a resistance box that is calibrated less than once/year, and can't provide the uncertainties, and as far as I can tell (although you disagree), it would appear the nominal value of the resistors are used. . It does not exactly inspire a lot of confidence. Dave
FT
Florian Teply
Tue, Apr 24, 2018 4:36 AM

Am Mon, 23 Apr 2018 23:59:57 +0100
schrieb "Dr. David Kirkby" drkirkby@kirkbymicrowave.co.uk:

On 21 April 2018 at 09:32, Florian Teply usenet@teply.info wrote:

Am Fri, 20 Apr 2018 14:37:22 +0100
schrieb "Dr. David Kirkby" drkirkby@kirkbymicrowave.co.uk:

The columns below, from left to right are

Device type (whether the DUT is floating, or grounded one side).
Resistor setting (ohms)
Votage (V)
Measurement time (Long or Short)
Test limits (+/- ohm)
Test results (ohm)

FLOAT 1E6 100 SHORT +/- 0.0086E6 -.0019E6
FLOAT 1E7 100 LONG +/- 0.0063E7 -.0016E7
FLOAT 1E8 100 LONG +/- 0.0073E8 -.0027E8
FLOAT 1E9 100 LONG +/- 0.0093E9 -.0032E9
FLOAT 1E10 100 LONG +/- 0.0273E10 +.0095E10
FLOAT 1E11 100 LONG +/- 0.0453E11 +.0080E11
FLOAT 1E11 100 SHORT +/- 0.0550E11 +.0086E11
FLOAT 1E11 10 LONG +/- 0.0546E11 +.0113E11
GROUND 1E7 100 LONG +/- 0.0065E7 -.0017E7
GROUND 1E11 10 LONG +/- 0.0573E11 +.0107E11

That strikes me that the assumption is the values are what their
nominal values are, but I wonder how accurate they are.

I might be wrong, but to me it seems like the resistors are not
exactly nominal but slightly off. But the uncertainty of the
measurements is larger than the deviation. Or were you referring to
the notion that the uncertainties are symmetrically distributed?
I'd be pretty surprised if the uncertainties were asymmetrical for
that matter.

how do you determine that the resistors are not assumed to be the
nominal value?

As far as I can see, taking the example of a 1e11 ohm resistor
grounded at one end (very last entry on table), the meter should read
1e11 +/- 0.0573e11 ohms. My meter read 0.010e11 ohms high, so was in
spec, as 0.010e11 is less than 0.0573e11. As far as I can determine,
the fact the permissable range of the meter is +/-x, rather than +x,
-y, means the nominal values are assumed.

Ah, probably I didn't get my wording precise. Just as you say, it looks
like they assume the resistors should have nominal value. Otherwise
they would need to list what they assume to be the nominal value, which
they don't. I was referring to the last column where they list the
measured deviation of +0.0107e11. In any case, I would consider unknown
resistances pretty odd as that would render the whole effort of
having 8.5 digits in the first place useless... So the only
explanation that would make sense is that the resistors should habve
nominal values. Unless of course the have individual values from
manufacture stored somewhere. But then it also wouldn't make sense to
not tell the owner of the device...

They don't spell out explicitlyy whether the test limits they give,
i.e. here +/- 0.0573e11 ohms would be their own measurement
uncvertainty or the instruments spec limits, but given the odd numbers
I'd expect it to be their measurement uncertainty.

Best regards,
Florian

Am Mon, 23 Apr 2018 23:59:57 +0100 schrieb "Dr. David Kirkby" <drkirkby@kirkbymicrowave.co.uk>: > On 21 April 2018 at 09:32, Florian Teply <usenet@teply.info> wrote: > > > Am Fri, 20 Apr 2018 14:37:22 +0100 > > schrieb "Dr. David Kirkby" <drkirkby@kirkbymicrowave.co.uk>: > > > > > The columns below, from left to right are > > > > > > Device type (whether the DUT is floating, or grounded one side). > > > Resistor setting (ohms) > > > Votage (V) > > > Measurement time (Long or Short) > > > Test limits (+/- ohm) > > > Test results (ohm) > > > > > > FLOAT 1E6 100 SHORT +/- 0.0086E6 -.0019E6 > > > FLOAT 1E7 100 LONG +/- 0.0063E7 -.0016E7 > > > FLOAT 1E8 100 LONG +/- 0.0073E8 -.0027E8 > > > FLOAT 1E9 100 LONG +/- 0.0093E9 -.0032E9 > > > FLOAT 1E10 100 LONG +/- 0.0273E10 +.0095E10 > > > FLOAT 1E11 100 LONG +/- 0.0453E11 +.0080E11 > > > FLOAT 1E11 100 SHORT +/- 0.0550E11 +.0086E11 > > > FLOAT 1E11 10 LONG +/- 0.0546E11 +.0113E11 > > > GROUND 1E7 100 LONG +/- 0.0065E7 -.0017E7 > > > GROUND 1E11 10 LONG +/- 0.0573E11 +.0107E11 > > > > > > That strikes me that the assumption is the values are what their > > > nominal values are, but I wonder how accurate they are. > > > > > I might be wrong, but to me it seems like the resistors are not > > exactly nominal but slightly off. But the uncertainty of the > > measurements is larger than the deviation. Or were you referring to > > the notion that the uncertainties are symmetrically distributed? > > I'd be pretty surprised if the uncertainties were asymmetrical for > > that matter. > > > how do you determine that the resistors are not assumed to be the > nominal value? > > As far as I can see, taking the example of a 1e11 ohm resistor > grounded at one end (very last entry on table), the meter should read > 1e11 +/- 0.0573e11 ohms. My meter read 0.010e11 ohms high, so was in > spec, as 0.010e11 is less than 0.0573e11. As far as I can determine, > the fact the permissable range of the meter is +/-x, rather than +x, > -y, means the nominal values are assumed. > Ah, probably I didn't get my wording precise. Just as you say, it looks like they assume the resistors should have nominal value. Otherwise they would need to list what they assume to be the nominal value, which they don't. I was referring to the last column where they list the measured deviation of +0.0107e11. In any case, I would consider unknown resistances pretty odd as that would render the whole effort of having 8.5 digits in the first place useless... So the only explanation that would make sense is that the resistors should habve nominal values. Unless of course the have individual values from manufacture stored somewhere. But then it also wouldn't make sense to not tell the owner of the device... They don't spell out explicitlyy whether the test limits they give, i.e. here +/- 0.0573e11 ohms would be their own measurement uncvertainty or the instruments spec limits, but given the odd numbers I'd expect it to be their measurement uncertainty. Best regards, Florian
DD
Dr. David Kirkby
Tue, Apr 24, 2018 12:19 PM

On Tue, 24 Apr 2018, 05:37 Florian Teply, usenet@teply.info wrote:

probably I didn't get my wording precise. Just as you say, it looks
like they assume the resistors should have nominal value. Otherwise
they would need to list what they assume to be the nominal value, which
they don't. I was referring to the last column where they list the
measured deviation of +0.0107e11. In any case, I would consider unknown
resistances pretty odd as that would render the whole effort of
having 8.5 digits in the first place useless... So the only
explanation that would make sense is that the resistors should habve
nominal values. Unless of course the have individual values from
manufacture stored somewhere. But then it also wouldn't make sense to
not tell the owner of the device...

The meter sent for calibration was a 4339B. That's a 5.5 digit high
resistance, not an 8.5 digit one.

They don't spell out explicitlyy whether the test limits they give,
i.e. here +/- 0.0573e11 ohms would be their own measurement
uncvertainty or the instruments spec limits, but given the odd numbers
I'd expect it to be their measurement uncertainty.

Best regards,
Florian

My interpretation is the numbers in the last column are the difference
between what Keysight believe is the correct value and what my meter read.
If you look on page 5, on Current Measurement Accuracy Test, you will see
for the 10 nA range, you will see test limits of +/- 0.063 nA and test
results of -.082 nA - so a failure.

The meter is going off to Keysight (UK) tomorrow. They may ore many not
send it to the USA, as they are waiting to hear from the USA whether they
can provide uncertainties.

Dave

On Tue, 24 Apr 2018, 05:37 Florian Teply, <usenet@teply.info> wrote: > probably I didn't get my wording precise. Just as you say, it looks > like they assume the resistors should have nominal value. Otherwise > they would need to list what they assume to be the nominal value, which > they don't. I was referring to the last column where they list the > measured deviation of +0.0107e11. In any case, I would consider unknown > resistances pretty odd as that would render the whole effort of > having 8.5 digits in the first place useless... So the only > explanation that would make sense is that the resistors should habve > nominal values. Unless of course the have individual values from > manufacture stored somewhere. But then it also wouldn't make sense to > not tell the owner of the device... > The meter sent for calibration was a 4339B. That's a 5.5 digit high resistance, not an 8.5 digit one. > > They don't spell out explicitlyy whether the test limits they give, > i.e. here +/- 0.0573e11 ohms would be their own measurement > uncvertainty or the instruments spec limits, but given the odd numbers > I'd expect it to be their measurement uncertainty. > > Best regards, > Florian My interpretation is the numbers in the last column are the *difference* between what Keysight believe is the correct value and what my meter read. If you look on page 5, on Current Measurement Accuracy Test, you will see for the 10 nA range, you will see test limits of +/- 0.063 nA and test results of -.082 nA - so a failure. The meter is going off to Keysight (UK) tomorrow. They may ore many not send it to the USA, as they are waiting to hear from the USA whether they can provide uncertainties. Dave