The -12V is used to drive the RS-232 signals... it also eventually gets to the EFC dac so it can swing below ground.  Also, the +12V gets to the DAC.  Pay close attention to generating noise on these lines.

Also, when I did Lady Heather's temperature control PID (with great help from Warren Sarkison) I did some  measurements on the effects of temperature on the system with only the tbolt under temperature control and also with both the power supply (the brick supplied with the original TAPR orders) and the tbolt temperature stabilized.  It seems about 1/4 to 1/3  of the temperature sensitivity of the system could be attributed to the power supply and the rest came from the oscillator/tbolt.  Some attention to the tempco of the power supply would be a good thing.

The -12V is used to drive the RS-232 signals... it also eventually gets to the EFC dac so it can swing below ground. Also, the +12V gets to the DAC. Pay close attention to generating noise on these lines. Also, when I did Lady Heather's temperature control PID (with great help from Warren Sarkison) I did some measurements on the effects of temperature on the system with only the tbolt under temperature control and also with both the power supply (the brick supplied with the original TAPR orders) and the tbolt temperature stabilized. It seems about 1/4 to 1/3 of the temperature sensitivity of the system could be attributed to the power supply and the rest came from the oscillator/tbolt. Some attention to the tempco of the power supply would be a good thing.
NS
Nick Sayer
Wed, Aug 31, 2016 5:56 AM

On Aug 30, 2016, at 9:26 PM, Mark Sims holrum@hotmail.com wrote:

The -12V is used to drive the RS-232 signals... it also eventually gets to the EFC dac so it can swing below ground.  Also, the +12V gets to the DAC.  Pay close attention to generating noise on these lines.

I’ve designed a prototype. I’ll provide noise and ripple measurements when I get the first boards back. I’m going to shoot for <= 35 mV P-P. If you need better than that, then it probably turns into a hybrid switching+linear system, which would be more expensive. It’s been a while since I looked, but I believe that’s what I was able to achieve with the supplies for the FE-5680A GPS discipline board.

Also, when I did Lady Heather's temperature control PID (with great help from Warren Sarkison) I did some  measurements on the effects of temperature on the system with only the tbolt under temperature control and also with both the power supply (the brick supplied with the original TAPR orders) and the tbolt temperature stabilized.  It seems about 1/4 to 1/3  of the temperature sensitivity of the system could be attributed to the power supply and the rest came from the oscillator/tbolt.  Some attention to the tempco of the power supply would be a good thing.

I’ve not made any efforts along those lines to start with. For one, the input is +12 V and is simply passed through as the +12 volt output (there is a TVS and cap on the input, however). Given that the Thunderbolt that I have on loan came with what I can only guess was once a laptop power supply, I started off thinking the bar was pretty low.

If you want the hybrid approach, then adding the temperature compensation to the linear regulators would clearly be the way to go.

> On Aug 30, 2016, at 9:26 PM, Mark Sims <holrum@hotmail.com> wrote: > > The -12V is used to drive the RS-232 signals... it also eventually gets to the EFC dac so it can swing below ground. Also, the +12V gets to the DAC. Pay close attention to generating noise on these lines. I’ve designed a prototype. I’ll provide noise and ripple measurements when I get the first boards back. I’m going to shoot for <= 35 mV P-P. If you need better than that, then it probably turns into a hybrid switching+linear system, which would be more expensive. It’s been a while since I looked, but I believe that’s what I was able to achieve with the supplies for the FE-5680A GPS discipline board. > > Also, when I did Lady Heather's temperature control PID (with great help from Warren Sarkison) I did some measurements on the effects of temperature on the system with only the tbolt under temperature control and also with both the power supply (the brick supplied with the original TAPR orders) and the tbolt temperature stabilized. It seems about 1/4 to 1/3 of the temperature sensitivity of the system could be attributed to the power supply and the rest came from the oscillator/tbolt. Some attention to the tempco of the power supply would be a good thing. I’ve not made any efforts along those lines to start with. For one, the input is +12 V and is simply passed through as the +12 volt output (there is a TVS and cap on the input, however). Given that the Thunderbolt that I have on loan came with what I can only guess was once a laptop power supply, I started off thinking the bar was pretty low. If you want the hybrid approach, then adding the temperature compensation to the linear regulators would clearly be the way to go.
CS
Charles Steinmetz
Wed, Aug 31, 2016 8:48 AM

Nick wrote:>

I’m going to shoot for <= 35 mV P-P. If you need better than that, then it probably turns into a hybrid switching+linear system

It's been years since I played with Tbolts (as opposed to just using
them), but as I recall there was no possibility whatever of getting the
best results from any switching supply without at least one stage of
linear "polishing."  In fact, everything I tried with initial switching
regulation benefitted from two levels of linear polishing.  It was my
conclusion that starting with a switching supply of any sort is a bad idea.

I found that for best performance the -12v supply, which feeds the DAC,
needs to be about 1000 times better than what you are shooting for
(i.e., low tens of uVp-p).  That surprised me, because I assumed there
must be some internal regulation of the DAC supplies -- but it was what
my testing showed.

Sure, a Tbolt will work with dirtier power than that, but for real
time-nuts performance I consider an all-linear supply with an
ultra-low-noise -12v rail to be absolutely necessary.  Since that is not
very hard to provide, I see no reason to try to make do with less.

One low-cost solution is to wait for a cheap Power One HTAA-16W-A
triple-output power supply to show up on ebay, and do a very minor
modification to raise the +12v current limit (so it will handle the
first moments of a cold start without current limiting) as detailed here:

http://www.ko4bb.com/getsimple/index.php?id=download&file=06_Misc_Test_Equipment/Power_One/Power_One_HTAA-16W-A_mod_for_Thunderbolt_power_supply.pdf

That approach achieves excellent results -- close to the best you can do
-- with very little fiddling.  If one wants the very best performance,
the -12v supply can be adjusted down to -16v or so to provide the
headroom to add a ULN regulator or a capacitance multiplier.  (The Tbolt
only draws a few mA from the -12v supply, so the ULN regulator can be a
suitable op amp with no external pass transistors.)

Last time I bought them, I paid $25 for a lot of three HTAA-16W-A's.

Best regards,

Charles

Nick wrote:> > I’m going to shoot for <= 35 mV P-P. If you need better than that, then it probably turns into a hybrid switching+linear system It's been years since I played with Tbolts (as opposed to just using them), but as I recall there was no possibility whatever of getting the best results from any switching supply without at least one stage of linear "polishing." In fact, everything I tried with initial switching regulation benefitted from two levels of linear polishing. It was my conclusion that starting with a switching supply of any sort is a bad idea. I found that for best performance the -12v supply, which feeds the DAC, needs to be about 1000 times better than what you are shooting for (i.e., low tens of uVp-p). That surprised me, because I assumed there must be some internal regulation of the DAC supplies -- but it was what my testing showed. Sure, a Tbolt will work with dirtier power than that, but for real time-nuts performance I consider an all-linear supply with an ultra-low-noise -12v rail to be absolutely necessary. Since that is not very hard to provide, I see no reason to try to make do with less. One low-cost solution is to wait for a cheap Power One HTAA-16W-A triple-output power supply to show up on ebay, and do a very minor modification to raise the +12v current limit (so it will handle the first moments of a cold start without current limiting) as detailed here: <http://www.ko4bb.com/getsimple/index.php?id=download&file=06_Misc_Test_Equipment/Power_One/Power_One_HTAA-16W-A_mod_for_Thunderbolt_power_supply.pdf> That approach achieves excellent results -- close to the best you can do -- with very little fiddling. If one wants the very best performance, the -12v supply can be adjusted down to -16v or so to provide the headroom to add a ULN regulator or a capacitance multiplier. (The Tbolt only draws a few mA from the -12v supply, so the ULN regulator can be a suitable op amp with no external pass transistors.) Last time I bought them, I paid $25 for a lot of three HTAA-16W-A's. Best regards, Charles
JO
Jerry O. Stern
Wed, Aug 31, 2016 11:57 AM

Charles, do you have the schematic for the HTAA-16W-A  ?  The KO4BB mod has limited info

Tnx

Jerry
K1jos@arrl.net

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Charles Steinmetz
Sent: Wednesday, August 31, 2016 4:49 AM
To: time-nuts@febo.com
Subject: Re: [time-nuts] Anybody want a Thunderbolt power supply?

Nick wrote:>

I’m going to shoot for <= 35 mV P-P. If you need better than that,
then it probably turns into a hybrid switching+linear system

It's been years since I played with Tbolts (as opposed to just using them), but as I recall there was no possibility whatever of getting the best results from any switching supply without at least one stage of linear "polishing."  In fact, everything I tried with initial switching regulation benefitted from two levels of linear polishing.  It was my conclusion that starting with a switching supply of any sort is a bad idea.

I found that for best performance the -12v supply, which feeds the DAC, needs to be about 1000 times better than what you are shooting for (i.e., low tens of uVp-p).  That surprised me, because I assumed there must be some internal regulation of the DAC supplies -- but it was what my testing showed.

Sure, a Tbolt will work with dirtier power than that, but for real time-nuts performance I consider an all-linear supply with an ultra-low-noise -12v rail to be absolutely necessary.  Since that is not very hard to provide, I see no reason to try to make do with less.

One low-cost solution is to wait for a cheap Power One HTAA-16W-A triple-output power supply to show up on ebay, and do a very minor modification to raise the +12v current limit (so it will handle the first moments of a cold start without current limiting) as detailed here:

http://www.ko4bb.com/getsimple/index.php?id=download&file=06_Misc_Test_Equipment/Power_One/Power_One_HTAA-16W-A_mod_for_Thunderbolt_power_supply.pdf

That approach achieves excellent results -- close to the best you can do
-- with very little fiddling.  If one wants the very best performance, the -12v supply can be adjusted down to -16v or so to provide the headroom to add a ULN regulator or a capacitance multiplier.  (The Tbolt only draws a few mA from the -12v supply, so the ULN regulator can be a suitable op amp with no external pass transistors.)

Last time I bought them, I paid $25 for a lot of three HTAA-16W-A's.

Best regards,

Charles


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Charles, do you have the schematic for the HTAA-16W-A ? The KO4BB mod has limited info Tnx Jerry K1jos@arrl.net -----Original Message----- From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Charles Steinmetz Sent: Wednesday, August 31, 2016 4:49 AM To: time-nuts@febo.com Subject: Re: [time-nuts] Anybody want a Thunderbolt power supply? Nick wrote:> > I’m going to shoot for <= 35 mV P-P. If you need better than that, > then it probably turns into a hybrid switching+linear system It's been years since I played with Tbolts (as opposed to just using them), but as I recall there was no possibility whatever of getting the best results from any switching supply without at least one stage of linear "polishing." In fact, everything I tried with initial switching regulation benefitted from two levels of linear polishing. It was my conclusion that starting with a switching supply of any sort is a bad idea. I found that for best performance the -12v supply, which feeds the DAC, needs to be about 1000 times better than what you are shooting for (i.e., low tens of uVp-p). That surprised me, because I assumed there must be some internal regulation of the DAC supplies -- but it was what my testing showed. Sure, a Tbolt will work with dirtier power than that, but for real time-nuts performance I consider an all-linear supply with an ultra-low-noise -12v rail to be absolutely necessary. Since that is not very hard to provide, I see no reason to try to make do with less. One low-cost solution is to wait for a cheap Power One HTAA-16W-A triple-output power supply to show up on ebay, and do a very minor modification to raise the +12v current limit (so it will handle the first moments of a cold start without current limiting) as detailed here: <http://www.ko4bb.com/getsimple/index.php?id=download&file=06_Misc_Test_Equipment/Power_One/Power_One_HTAA-16W-A_mod_for_Thunderbolt_power_supply.pdf> That approach achieves excellent results -- close to the best you can do -- with very little fiddling. If one wants the very best performance, the -12v supply can be adjusted down to -16v or so to provide the headroom to add a ULN regulator or a capacitance multiplier. (The Tbolt only draws a few mA from the -12v supply, so the ULN regulator can be a suitable op amp with no external pass transistors.) Last time I bought them, I paid $25 for a lot of three HTAA-16W-A's. Best regards, Charles _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
CJ
Clint Jay
Wed, Aug 31, 2016 12:59 PM

Ah,  these are the LM723 based linear supplies,  Lambda Coutant made
variants as did Farnell,  they're renowned as high reliability,  low noise
supplies.

There are schematics out there for the variants but they're all extremely
similar if not identical.

I think I may even have a couple of original manuals with schematics,  I
may be able to dig them out and scan if they're of interest.

On 31 Aug 2016 13:01, "Jerry O. Stern" jsternmd@att.net wrote:

Charles, do you have the schematic for the HTAA-16W-A  ?  The KO4BB mod
has limited info

Tnx

Jerry
K1jos@arrl.net

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Charles
Steinmetz
Sent: Wednesday, August 31, 2016 4:49 AM
To: time-nuts@febo.com
Subject: Re: [time-nuts] Anybody want a Thunderbolt power supply?

Nick wrote:>

I’m going to shoot for <= 35 mV P-P. If you need better than that,
then it probably turns into a hybrid switching+linear system

It's been years since I played with Tbolts (as opposed to just using
them), but as I recall there was no possibility whatever of getting the
best results from any switching supply without at least one stage of linear
"polishing."  In fact, everything I tried with initial switching regulation
benefitted from two levels of linear polishing.  It was my conclusion that
starting with a switching supply of any sort is a bad idea.

I found that for best performance the -12v supply, which feeds the DAC,
needs to be about 1000 times better than what you are shooting for (i.e.,
low tens of uVp-p).  That surprised me, because I assumed there must be
some internal regulation of the DAC supplies -- but it was what my testing
showed.

Sure, a Tbolt will work with dirtier power than that, but for real
time-nuts performance I consider an all-linear supply with an
ultra-low-noise -12v rail to be absolutely necessary.  Since that is not
very hard to provide, I see no reason to try to make do with less.

One low-cost solution is to wait for a cheap Power One HTAA-16W-A
triple-output power supply to show up on ebay, and do a very minor
modification to raise the +12v current limit (so it will handle the first
moments of a cold start without current limiting) as detailed here:

<http://www.ko4bb.com/getsimple/index.php?id=download&file=06_Misc_Test_
Equipment/Power_One/Power_One_HTAA-16W-A_mod_for_
Thunderbolt_power_supply.pdf>

That approach achieves excellent results -- close to the best you can do
-- with very little fiddling.  If one wants the very best performance, the
-12v supply can be adjusted down to -16v or so to provide the headroom to
add a ULN regulator or a capacitance multiplier.  (The Tbolt only draws a
few mA from the -12v supply, so the ULN regulator can be a suitable op amp
with no external pass transistors.)

Last time I bought them, I paid $25 for a lot of three HTAA-16W-A's.

Best regards,

Charles


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

Ah, these are the LM723 based linear supplies, Lambda Coutant made variants as did Farnell, they're renowned as high reliability, low noise supplies. There are schematics out there for the variants but they're all extremely similar if not identical. I think I may even have a couple of original manuals with schematics, I may be able to dig them out and scan if they're of interest. On 31 Aug 2016 13:01, "Jerry O. Stern" <jsternmd@att.net> wrote: > Charles, do you have the schematic for the HTAA-16W-A ? The KO4BB mod > has limited info > > Tnx > > Jerry > K1jos@arrl.net > > > -----Original Message----- > From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Charles > Steinmetz > Sent: Wednesday, August 31, 2016 4:49 AM > To: time-nuts@febo.com > Subject: Re: [time-nuts] Anybody want a Thunderbolt power supply? > > Nick wrote:> > > > I’m going to shoot for <= 35 mV P-P. If you need better than that, > > then it probably turns into a hybrid switching+linear system > > It's been years since I played with Tbolts (as opposed to just using > them), but as I recall there was no possibility whatever of getting the > best results from any switching supply without at least one stage of linear > "polishing." In fact, everything I tried with initial switching regulation > benefitted from two levels of linear polishing. It was my conclusion that > starting with a switching supply of any sort is a bad idea. > > I found that for best performance the -12v supply, which feeds the DAC, > needs to be about 1000 times better than what you are shooting for (i.e., > low tens of uVp-p). That surprised me, because I assumed there must be > some internal regulation of the DAC supplies -- but it was what my testing > showed. > > Sure, a Tbolt will work with dirtier power than that, but for real > time-nuts performance I consider an all-linear supply with an > ultra-low-noise -12v rail to be absolutely necessary. Since that is not > very hard to provide, I see no reason to try to make do with less. > > One low-cost solution is to wait for a cheap Power One HTAA-16W-A > triple-output power supply to show up on ebay, and do a very minor > modification to raise the +12v current limit (so it will handle the first > moments of a cold start without current limiting) as detailed here: > > <http://www.ko4bb.com/getsimple/index.php?id=download&file=06_Misc_Test_ > Equipment/Power_One/Power_One_HTAA-16W-A_mod_for_ > Thunderbolt_power_supply.pdf> > > That approach achieves excellent results -- close to the best you can do > -- with very little fiddling. If one wants the very best performance, the > -12v supply can be adjusted down to -16v or so to provide the headroom to > add a ULN regulator or a capacitance multiplier. (The Tbolt only draws a > few mA from the -12v supply, so the ULN regulator can be a suitable op amp > with no external pass transistors.) > > Last time I bought them, I paid $25 for a lot of three HTAA-16W-A's. > > Best regards, > > Charles > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > and follow the instructions there. >
CS
Charles Steinmetz
Wed, Aug 31, 2016 6:22 PM

Jerry wrote:

Charles, do you have the schematic for the HTAA-16W-A  ?  The KO4BB mod has limited info

Clint wrote:

Ah,  these are the LM723 based linear supplies,  Lambda Coutant made
variants as did Farnell,  they're renowned as high reliability,  low

noise

supplies.

There are schematics out there for the variants but they're all

extremely

similar if not identical.

I don't have a schematic for the HTAA-16W-A, but Clint is correct --
each of the three supplies is a bog-standard LM723 circuit.  I am
attaching a reverse-engineered schematic I found on the net of a
different Power One supply as an example.  I hope I've made it small
enough to get through the server without list-minder approval, in case
Tom is still away.

Best regards,

Charles

Jerry wrote: > Charles, do you have the schematic for the HTAA-16W-A ? The KO4BB mod has limited info Clint wrote: >> Ah, these are the LM723 based linear supplies, Lambda Coutant made >> variants as did Farnell, they're renowned as high reliability, low noise >> supplies. >> >> There are schematics out there for the variants but they're all extremely >> similar if not identical. I don't have a schematic for the HTAA-16W-A, but Clint is correct -- each of the three supplies is a bog-standard LM723 circuit. I am attaching a reverse-engineered schematic I found on the net of a different Power One supply as an example. I hope I've made it small enough to get through the server without list-minder approval, in case Tom is still away. Best regards, Charles
J
jimlux
Wed, Aug 31, 2016 7:26 PM

On 8/31/16 11:22 AM, Charles Steinmetz wrote:

Jerry wrote:

Charles, do you have the schematic for the HTAA-16W-A  ?  The KO4BB
mod has limited info

Clint wrote:

Ah,  these are the LM723 based linear supplies,  Lambda Coutant made
variants as did Farnell,  they're renowned as high reliability,  low

noise

supplies.

There are schematics out there for the variants but they're all

extremely

similar if not identical.

I don't have a schematic for the HTAA-16W-A, but Clint is correct --
each of the three supplies is a bog-standard LM723 circuit.

with the venerable 2n3055 (1960s!) as the series pass device, and an
exotic TIP29 (my 1985 databook has it, but the original datasheet says
1968) to drive it..

They probably used the 1970s versions of the transistor, of course,
since the 723 didn't come out til early 70s..

Some things just keep working...
I think they've amply recovered the development cost for those parts..

I am

attaching a reverse-engineered schematic I found on the net of a
different Power One supply as an example.  I hope I've made it small
enough to get through the server without list-minder approval, in case
Tom is still away.

Best regards,

Charles


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

On 8/31/16 11:22 AM, Charles Steinmetz wrote: > Jerry wrote: > >> Charles, do you have the schematic for the HTAA-16W-A ? The KO4BB >> mod has limited info > > Clint wrote: > >>> Ah, these are the LM723 based linear supplies, Lambda Coutant made >>> variants as did Farnell, they're renowned as high reliability, low > noise >>> supplies. >>> >>> There are schematics out there for the variants but they're all > extremely >>> similar if not identical. > > I don't have a schematic for the HTAA-16W-A, but Clint is correct -- > each of the three supplies is a bog-standard LM723 circuit. with the venerable 2n3055 (1960s!) as the series pass device, and an exotic TIP29 (my 1985 databook has it, but the original datasheet says 1968) to drive it.. They probably used the 1970s versions of the transistor, of course, since the 723 didn't come out til early 70s.. Some things just keep working... I think they've amply recovered the development cost for those parts.. I am > attaching a reverse-engineered schematic I found on the net of a > different Power One supply as an example. I hope I've made it small > enough to get through the server without list-minder approval, in case > Tom is still away. > > Best regards, > > Charles > > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
CS
Charles Steinmetz
Wed, Aug 31, 2016 9:12 PM

jim wrote:

with the venerable 2n3055 (1960s!) as the series pass device, and an
exotic TIP29 (my 1985 databook has it, but the original datasheet says
1968) to drive it..

They probably used the 1970s versions of the transistor, of course,
since the 723 didn't come out til early 70s..

Fairchild introduced the orginal 723 (the µA723) in 1968.  I think
National brought out the LM723 within the year, but can't swear to it.

While the schematic I posted shows a TIP29 driver transistor, that was
for a different, 3A supply.  The lower-current +/-12v supplies in the
HTAA-16W-A do without the driver, and connect the 3055 directly to the
723 output (which can source up to 150mA).  The schematic I posted also
shows foldback current limiting, which the +/-12v supplies in the
HTAA-16W-A do not have.

Best regards,

Charles

jim wrote: > with the venerable 2n3055 (1960s!) as the series pass device, and an > exotic TIP29 (my 1985 databook has it, but the original datasheet says > 1968) to drive it.. > > They probably used the 1970s versions of the transistor, of course, > since the 723 didn't come out til early 70s.. Fairchild introduced the orginal 723 (the µA723) in 1968. I think National brought out the LM723 within the year, but can't swear to it. While the schematic I posted shows a TIP29 driver transistor, that was for a different, 3A supply. The lower-current +/-12v supplies in the HTAA-16W-A do without the driver, and connect the 3055 directly to the 723 output (which can source up to 150mA). The schematic I posted also shows foldback current limiting, which the +/-12v supplies in the HTAA-16W-A do not have. Best regards, Charles