question about HP 5601a harmonic generator

C
cdelect@juno.com
Mon, Dec 25, 2017 6:53 PM

When working on the A4 assy. the ONLY adjustments you should make are:

  • Cavity frequency, unscrew locking ring and while looking at the A7 J1
    signal or 2nd harmonic reading adjust for max and then lock down.

  • Attenuator, Unlock with allen, adjust for same, usually pulled all the
    way out. Remember to lock.

-S5 the one by the attenuator, only needed if tube swapped but if you
adjust make sure the peak does not slip when you tighten the locknut.

  • the top pot, adjust for the same.

  • If you adjust any of the other locknutted plungers you will NEVER get
    back to where you need to be!!!!

Cheers,

Corby

When working on the A4 assy. the ONLY adjustments you should make are: - Cavity frequency, unscrew locking ring and while looking at the A7 J1 signal or 2nd harmonic reading adjust for max and then lock down. - Attenuator, Unlock with allen, adjust for same, usually pulled all the way out. Remember to lock. -S5 the one by the attenuator, only needed if tube swapped but if you adjust make sure the peak does not slip when you tighten the locknut. - the top pot, adjust for the same. - If you adjust any of the other locknutted plungers you will NEVER get back to where you need to be!!!! Cheers, Corby