HP5065A C-field mods and optical unit mods
The contribution of the temperature coefficients of the fixed resistors
and the pot (which is a 21 turn bulk metal foil 10PPM/Deg C pot) show up
as frequency errors as shown for several combinations.
Note that the pot does not cover the entire 2X10-9 frequency range.
There is a high, mid, and low switch setting that shorts both 1k,
one 1k, or neither 1k resistor that in conjunction with the pot allows
the entire range to be covered. This allows the pot to have a finer
frequency resolution
A- With only the 2.4K in circuit with the pot at zero Ohms a 1 Deg C
change
will cause a 2.8X10-15 frequency change.
B- With the 2.4K in combination with the pot at 1K a 1 Deg C change will
cause a 8.55X10-15 frequency change.
C- With the 2.4K in combination with the pot at 1K and both 1K resistors
in
circuit a 1 Deg C change will cause a 3.99X10-15 frequency change.
If in example C you replace the pot with a 100PPM pot a 1 Deg C change
will equate to a 2.45X10-14 frequency change.
As you can see this circuit will work quite well in eliminating any power
supply related C-field changes.
Now the C-field coil is only partially temperature regulated by the
ovens in the optical unit as they regulate the lamp and cell
temperatures
and the C-field is not co-located.
That is why I'm going to use the PC liquid cooler to actively regulate
the
temperature of the sealed enclosure I have placed the optical unit into.
Cheers,
Corby
Hi
One interesting claim made “back in the good old days” was that certain
pots had near zero TC when used as radiometric devices. Essentially the
metal on one side “always” had the same characteristics as the metal on
the other side of the wiper. People would go to great extremes to exploit
this “well known fact”. Oddly enough I never really saw much proof that
this was true ….You still see people doing things this way though. It was
a major “design item” with a certain company that made both potentiometers
and OCXO’s ….
=======
At some point, the magnetic field of the car in the garage gets into
the act on your Rb standard. There is a point where all sorts of
things start to be questioned. Shielding is only just so good. That’s
true if it’s heat shielding, magnetic shielding or whatever ….
Once you get into the “part in 10^-15” range on an Rb, you are well past
the “show me” level of effects. You can (correctly) calculate that this or
that is the impact. Demonstrating that it works that way at that level
… not so much.
Why worry about demonstrating this or that? Often second or third order
effects pop up as you “null out” the first order stuff. Hysteresis in the TC
of a “compensated” standard is one very common example. At this point
it’s not a big surprise when it shows up. That was not always the case ….
None of this is easy, there always is a lot more work involved that one might
think.
Bob
On Nov 22, 2017, at 12:04 PM, cdelect@juno.com wrote:
HP5065A C-field mods and optical unit mods
The contribution of the temperature coefficients of the fixed resistors
and the pot (which is a 21 turn bulk metal foil 10PPM/Deg C pot) show up
as frequency errors as shown for several combinations.
Note that the pot does not cover the entire 2X10-9 frequency range.
There is a high, mid, and low switch setting that shorts both 1k,
one 1k, or neither 1k resistor that in conjunction with the pot allows
the entire range to be covered. This allows the pot to have a finer
frequency resolution
A- With only the 2.4K in circuit with the pot at zero Ohms a 1 Deg C
change
will cause a 2.8X10-15 frequency change.
B- With the 2.4K in combination with the pot at 1K a 1 Deg C change will
cause a 8.55X10-15 frequency change.
C- With the 2.4K in combination with the pot at 1K and both 1K resistors
in
circuit a 1 Deg C change will cause a 3.99X10-15 frequency change.
If in example C you replace the pot with a 100PPM pot a 1 Deg C change
will equate to a 2.45X10-14 frequency change.
As you can see this circuit will work quite well in eliminating any power
supply related C-field changes.
Now the C-field coil is only partially temperature regulated by the
ovens in the optical unit as they regulate the lamp and cell
temperatures
and the C-field is not co-located.
That is why I'm going to use the PC liquid cooler to actively regulate
the
temperature of the sealed enclosure I have placed the optical unit into.
Cheers,
Corby
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.