Mon, Jan 16, 2017 8:58 PM

I've been doing some work on testing some HP 5061A/B tubes
and came across something that's not making any sense!
The R11 and R12 values marked on the tubes for use with the
new style A11 are supposed to reflect a parallel value
that is 10X larger than the thermistor reading.
This to agree with the ratio of the bridge resistors in the
A11 module.
For instance a 3112A prefix tube shows values of 3.83K
and 42.2K. This gives a parallel value of 3.511K which when
divided by 10 gives 351.1 Ohm. The value marked on the tube
was 352 Ohm. That all makes sense!
Now I have three 3232A prefix tubes and they all do not
reflect this ratio.
For instance one has values of 51.1K and 5.11K which gives
4.65K. This divided by ten gives 465 Ohms. However the
Thermistor value marked on the tube is 915 Ohms!
I tested all three tubes and they come up to the proper
temperature with the thermistor value marked on the tube.
Did HP make a change to the A11 bridge resistor ratio, and
if so where did they document it!
If not and you used the R11 and R12 values you would
seriously overheat the tube!

Cheers,

Corby

I've been doing some work on testing some HP 5061A/B tubes and came across something that's not making any sense! The R11 and R12 values marked on the tubes for use with the new style A11 are supposed to reflect a parallel value that is 10X larger than the thermistor reading. This to agree with the ratio of the bridge resistors in the A11 module. For instance a 3112A prefix tube shows values of 3.83K and 42.2K. This gives a parallel value of 3.511K which when divided by 10 gives 351.1 Ohm. The value marked on the tube was 352 Ohm. That all makes sense! Now I have three 3232A prefix tubes and they all do not reflect this ratio. For instance one has values of 51.1K and 5.11K which gives 4.65K. This divided by ten gives 465 Ohms. However the Thermistor value marked on the tube is 915 Ohms! I tested all three tubes and they come up to the proper temperature with the thermistor value marked on the tube. Did HP make a change to the A11 bridge resistor ratio, and if so where did they document it! If not and you used the R11 and R12 values you would seriously overheat the tube! Cheers, Corby
PS
paul swed
Mon, Jan 16, 2017 10:07 PM

Corby thats a heck of a puzzle. I am trying to think about how you could
prove it. I have some thoughts that are not at all well formed. It comes
from my experiments on Frankenstein temp control.

If you could please tell me what you might expect the alternate bridge
resistors to be overall I will look through my various bits to see if there
is anything.

Regards
Paul
WB8TSL

On Mon, Jan 16, 2017 at 3:58 PM, cdelect@juno.com wrote:

I've been doing some work on testing some HP 5061A/B tubes
and came across something that's not making any sense!
The R11 and R12 values marked on the tubes for use with the
new style A11 are supposed to reflect a parallel value
that is 10X larger than the thermistor reading.
This to agree with the ratio of the bridge resistors in the
A11 module.
For instance a 3112A prefix tube shows values of 3.83K
and 42.2K. This gives a parallel value of 3.511K which when
divided by 10 gives 351.1 Ohm. The value marked on the tube
was 352 Ohm. That all makes sense!
Now I have three 3232A prefix tubes and they all do not
reflect this ratio.
For instance one has values of 51.1K and 5.11K which gives
4.65K. This divided by ten gives 465 Ohms. However the
Thermistor value marked on the tube is 915 Ohms!
I tested all three tubes and they come up to the proper
temperature with the thermistor value marked on the tube.
Did HP make a change to the A11 bridge resistor ratio, and
if so where did they document it!
If not and you used the R11 and R12 values you would
seriously overheat the tube!

Cheers,

Corby


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

Corby thats a heck of a puzzle. I am trying to think about how you could prove it. I have some thoughts that are not at all well formed. It comes from my experiments on Frankenstein temp control. If you could please tell me what you might expect the alternate bridge resistors to be overall I will look through my various bits to see if there is anything. Regards Paul WB8TSL On Mon, Jan 16, 2017 at 3:58 PM, <cdelect@juno.com> wrote: > I've been doing some work on testing some HP 5061A/B tubes > and came across something that's not making any sense! > The R11 and R12 values marked on the tubes for use with the > new style A11 are supposed to reflect a parallel value > that is 10X larger than the thermistor reading. > This to agree with the ratio of the bridge resistors in the > A11 module. > For instance a 3112A prefix tube shows values of 3.83K > and 42.2K. This gives a parallel value of 3.511K which when > divided by 10 gives 351.1 Ohm. The value marked on the tube > was 352 Ohm. That all makes sense! > Now I have three 3232A prefix tubes and they all do not > reflect this ratio. > For instance one has values of 51.1K and 5.11K which gives > 4.65K. This divided by ten gives 465 Ohms. However the > Thermistor value marked on the tube is 915 Ohms! > I tested all three tubes and they come up to the proper > temperature with the thermistor value marked on the tube. > Did HP make a change to the A11 bridge resistor ratio, and > if so where did they document it! > If not and you used the R11 and R12 values you would > seriously overheat the tube! > > Cheers, > > Corby > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > and follow the instructions there. >
PS
paul swed
Mon, Jan 23, 2017 1:22 AM

Corby
Curious if you ever received any other comments or had success?
Regards
Paul
WB8TSL

On Mon, Jan 16, 2017 at 5:07 PM, paul swed paulswedb@gmail.com wrote:

Corby thats a heck of a puzzle. I am trying to think about how you could
prove it. I have some thoughts that are not at all well formed. It comes
from my experiments on Frankenstein temp control.

If you could please tell me what you might expect the alternate bridge
resistors to be overall I will look through my various bits to see if there
is anything.

Regards
Paul
WB8TSL

On Mon, Jan 16, 2017 at 3:58 PM, cdelect@juno.com wrote:

I've been doing some work on testing some HP 5061A/B tubes
and came across something that's not making any sense!
The R11 and R12 values marked on the tubes for use with the
new style A11 are supposed to reflect a parallel value
that is 10X larger than the thermistor reading.
This to agree with the ratio of the bridge resistors in the
A11 module.
For instance a 3112A prefix tube shows values of 3.83K
and 42.2K. This gives a parallel value of 3.511K which when
divided by 10 gives 351.1 Ohm. The value marked on the tube
was 352 Ohm. That all makes sense!
Now I have three 3232A prefix tubes and they all do not
reflect this ratio.
For instance one has values of 51.1K and 5.11K which gives
4.65K. This divided by ten gives 465 Ohms. However the
Thermistor value marked on the tube is 915 Ohms!
I tested all three tubes and they come up to the proper
temperature with the thermistor value marked on the tube.
Did HP make a change to the A11 bridge resistor ratio, and
if so where did they document it!
If not and you used the R11 and R12 values you would
seriously overheat the tube!

Cheers,

Corby


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

Corby Curious if you ever received any other comments or had success? Regards Paul WB8TSL On Mon, Jan 16, 2017 at 5:07 PM, paul swed <paulswedb@gmail.com> wrote: > Corby thats a heck of a puzzle. I am trying to think about how you could > prove it. I have some thoughts that are not at all well formed. It comes > from my experiments on Frankenstein temp control. > > If you could please tell me what you might expect the alternate bridge > resistors to be overall I will look through my various bits to see if there > is anything. > > Regards > Paul > WB8TSL > > On Mon, Jan 16, 2017 at 3:58 PM, <cdelect@juno.com> wrote: > >> I've been doing some work on testing some HP 5061A/B tubes >> and came across something that's not making any sense! >> The R11 and R12 values marked on the tubes for use with the >> new style A11 are supposed to reflect a parallel value >> that is 10X larger than the thermistor reading. >> This to agree with the ratio of the bridge resistors in the >> A11 module. >> For instance a 3112A prefix tube shows values of 3.83K >> and 42.2K. This gives a parallel value of 3.511K which when >> divided by 10 gives 351.1 Ohm. The value marked on the tube >> was 352 Ohm. That all makes sense! >> Now I have three 3232A prefix tubes and they all do not >> reflect this ratio. >> For instance one has values of 51.1K and 5.11K which gives >> 4.65K. This divided by ten gives 465 Ohms. However the >> Thermistor value marked on the tube is 915 Ohms! >> I tested all three tubes and they come up to the proper >> temperature with the thermistor value marked on the tube. >> Did HP make a change to the A11 bridge resistor ratio, and >> if so where did they document it! >> If not and you used the R11 and R12 values you would >> seriously overheat the tube! >> >> Cheers, >> >> Corby >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/m >> ailman/listinfo/time-nuts >> and follow the instructions there. >> > >