Hi Miguel,
your S/N can be found in the Keysight database. Your instrument was sent
to customer on 15.12.2009, order number 2856337. No calibration
documents, or newer calibration data are available.
So it's now nearly 10 years old, and you should definitely save the
content of the CALRAM, and maybe already change the DALLAS chips. These
might last another 5 years, but nobody can make sure. If it has been
used for 7-8 years, maybe w/o calibration, its LTZ1000A reference has
already drifted a lot, so a calibration is required.. or you know some
volt-nuts who has access to a known / calibrated voltage reference to check.
You might also check CAL?72 over one week, according to SN18, which
requires constant room temperature, and a constant external reference to
check staibility of its internal LTZ reference.
All instruments might be affected to more or less extent by this A/D
converter drift problem (U180), so that's what Illya is referring to.
Also the LTZ1000A reference might need some time to recover, because it
might show some hysteresis or rush-in effect after 2 years storage.
If you want to get a proper calibration anyhow, you may also set the
oven temperature lower, to 65°C as explained, and then send it out for
calibration. Or if you already have a stable external reference, you can
preliminarily copy the calibration after changing the oven.
Frank
On Tue, 30 Jul 2019 at 07:03, Frank Stellmach frank.stellmach@freenet.de
wrote:
// My serial number is: MY45044988
Hi Miguel,
So it's now nearly 10 years old, and you should definitely save the
content of the CALRAM, and maybe already change the DALLAS chips. These
might last another 5 years, but nobody can make sure. If it has been
used for 7-8 years, maybe w/o calibration, its LTZ1000A reference has
already drifted a lot, so a calibration is required..
Also the LTZ1000A reference might need some time to recover, because it
might show some hysteresis or rush-in effect after 2 years storage.
How long should he run it continuously to recover? Before sending it for
calibration?
If you want to get a proper calibration anyhow, you may also set the
oven temperature lower, to 65°C as explained, and then send it out for
calibration. Or if you already have a stable external reference, you can
preliminarily copy the calibration after changing the oven.
How long should he run it continuously after changing the oven temperature?
I am guessing, (perhaps incorrectly), that there’s no point running it
continuously before changing the oven temperature if he intends changing
the oven temperature.
I’ve known Miguel for quite a few years - only by the Internet. I have
never met him in person. He doesn’t have access to any any stable
references, and I don’t think he would want to run air conditioning 24/7.
He can tell you how much the temperature varies throughout the year, if he
has logged it, which I doubt.
Being in Mexico, I don’t think there will be any local time nuts.
I personally think that he would be better getting the meter calibrated in
the USA rather than Mexico. He had been told it can be calibrated in
Mexico, but it would have to go to the USA to have the firmware upgraded!
From a number of things he has told me about Mexico over the years, I
personally would not want to get anything done there.
Frank
Dr. David Kirkby,
Running SN18 test for few weeks will show how stable is A3 ADC, and then it can be estimated if meter needs long running time. Waiting for weeks is typical when it comes to 8.5d DC measurements. There is no way around it , despite marketing efforts of various vendors. External reference is not required for SN18 test.
Also ignore test "limits" outlined in SN18 document, they are very wrong to judge stability of the ADC. Good meters must have drift less than 0.01 ppm/day, otherwise you just wasting money at expensive calibration which will be invalid before the next cycle arrives (assuming typical annual period). Good idea to keep ambient temperature stable (doesn't matter cold or hot, only deviation is important for SN18).
Running meter continuously reduce thermal cycling induced stress , at expense of wee bit larger drift per year. Lower temperature reference also take care of that.
On July 30, 2019 6:37:28 PM GMT+08:00, "Dr. David Kirkby" drkirkby@kirkbymicrowave.co.uk wrote:
On Tue, 30 Jul 2019 at 07:03, Frank Stellmach
frank.stellmach@freenet.de
wrote:
// My serial number is: MY45044988
Hi Miguel,
So it's now nearly 10 years old, and you should definitely save the
content of the CALRAM, and maybe already change the DALLAS chips.
These
might last another 5 years, but nobody can make sure. If it has been
used for 7-8 years, maybe w/o calibration, its LTZ1000A reference has
already drifted a lot, so a calibration is required..
Also the LTZ1000A reference might need some time to recover, because
it
might show some hysteresis or rush-in effect after 2 years storage.
How long should he run it continuously to recover? Before sending it
for
calibration?
If you want to get a proper calibration anyhow, you may also set the
oven temperature lower, to 65°C as explained, and then send it out
for
calibration. Or if you already have a stable external reference, you
can
preliminarily copy the calibration after changing the oven.
How long should he run it continuously after changing the oven
temperature?
I am guessing, (perhaps incorrectly), that there’s no point running it
continuously before changing the oven temperature if he intends
changing
the oven temperature.
I’ve known Miguel for quite a few years - only by the Internet. I have
never met him in person. He doesn’t have access to any any stable
references, and I don’t think he would want to run air conditioning
24/7.
He can tell you how much the temperature varies throughout the year, if
he
has logged it, which I doubt.
Being in Mexico, I don’t think there will be any local time nuts.
I personally think that he would be better getting the meter calibrated
in
the USA rather than Mexico. He had been told it can be calibrated in
Mexico, but it would have to go to the USA to have the firmware
upgraded!
From a number of things he has told me about Mexico over the years, I
personally would not want to get anything done there.
Frank
Dr. David Kirkby,
volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.
BR,
Illya Tsemenko
On 30/07/19 10:37 pm, Dr. David Kirkby wrote:
Being in Mexico, I don’t think there will be any local time nuts.
Well, I was in Guadalajara on and off between 1999 and 2012 after which I moved back to New Zealand. Unfortunately I wasn't on timenuts then so didn't know Miguel was there.
I personally think that he would be better getting the meter calibrated in
the USA rather than Mexico. He had been told it can be calibrated in
Mexico, but it would have to go to the USA to have the firmware upgraded!
From a number of things he has told me about Mexico over the years, I
personally would not want to get anything done there.
HP had a factory in GDL, Id have loved to been allowed to go dumpster diving there.