3458A question

DC
David C. Partridge
Thu, Jul 6, 2017 8:35 AM

After doing an ACAL, I got a reading of 1.0199986V on a voltage reference
set to 1.020000V

Over the next few minutes this edged upward and I'm now getting a reading
between 1.0199998 and 1.020000 which is what I'd expect.

Is what I'm seeing normal?

Dave

After doing an ACAL, I got a reading of 1.0199986V on a voltage reference set to 1.020000V Over the next few minutes this edged upward and I'm now getting a reading between 1.0199998 and 1.020000 which is what I'd expect. Is what I'm seeing normal? Dave
DC
David C. Partridge
Thu, Jul 6, 2017 8:47 AM

Hmmm - not using CuTe (TeCu) banana plugs, so it could just be thermals

Does anyone know where I can get some of these (and some spades) w/o paying Keysight a huge pile of Mars Bars

Dave
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of David C. Partridge
Sent: 06 July 2017 09:36
To: 'Discussion of precise voltage measurement'
Subject: [volt-nuts] 3458A question

After doing an ACAL, I got a reading of 1.0199986V on a voltage reference set to 1.020000V

Over the next few minutes this edged upward and I'm now getting a reading between 1.0199998 and 1.020000 which is what I'd expect.

Is what I'm seeing normal?

Dave


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Hmmm - not using CuTe (TeCu) banana plugs, so it could just be thermals Does anyone know where I can get some of these (and some spades) w/o paying Keysight a huge pile of Mars Bars Dave -----Original Message----- From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of David C. Partridge Sent: 06 July 2017 09:36 To: 'Discussion of precise voltage measurement' Subject: [volt-nuts] 3458A question After doing an ACAL, I got a reading of 1.0199986V on a voltage reference set to 1.020000V Over the next few minutes this edged upward and I'm now getting a reading between 1.0199998 and 1.020000 which is what I'd expect. Is what I'm seeing normal? Dave _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
DC
Domenico Crupi
Mon, Jul 10, 2017 11:07 AM

I noticed this behaviour after autocal also in my unit.
I think it may be caused by autocal procedure that exercise some relay
switch and this is resulting some uV offset due to EMF.
After waiting some time temperature will equalise again and accuracy will
be slight better.

I noticed this behaviour after autocal also in my unit. I think it may be caused by autocal procedure that exercise some relay switch and this is resulting some uV offset due to EMF. After waiting some time temperature will equalise again and accuracy will be slight better.