HP5328A + NI GPIB-ENET/1000

JP
James Peroulas
Thu, Feb 23, 2017 6:37 AM

Does anyone have experience using older HP equipment with the NI
GPIB-ENET/1000 GPIB to ethernet box? Is this even possible with such an old
piece of equipment? I'm very familiar with using SCPI commands over telnet
but not so familiar with HPIB.

So far I've installed the NI 488.2 software which has successfully found
both the GPIB-ENET/1000 box and the HP5328A connected to it. I've opened
the NI-488.2 communicator which should allow me to send commands to the
counter and read back results. However, anything I type in the "send
string" returns a timeout.

The manual for the 5328A shows example commands such as:
CMD"?U*","T"
But it's not clear to me what portion (all?) is the string that I should be
sending to the 5328a.

The 5328a is set to "Talk Only" mode via dip switch. I would expect that I
should be able to just 'read' the results, but 'read' also times out...

Thanks for any pointers,
James

Does anyone have experience using older HP equipment with the NI GPIB-ENET/1000 GPIB to ethernet box? Is this even possible with such an old piece of equipment? I'm very familiar with using SCPI commands over telnet but not so familiar with HPIB. So far I've installed the NI 488.2 software which has successfully found both the GPIB-ENET/1000 box and the HP5328A connected to it. I've opened the NI-488.2 communicator which should allow me to send commands to the counter and read back results. However, anything I type in the "send string" returns a timeout. The manual for the 5328A shows example commands such as: CMD"?U*","T" But it's not clear to me what portion (all?) is the string that I should be sending to the 5328a. The 5328a is set to "Talk Only" mode via dip switch. I would expect that I should be able to just 'read' the results, but 'read' also times out... Thanks for any pointers, James
SM
Scott McGrath
Thu, Feb 23, 2017 4:21 PM

'Talk Only' mode responds to serial poll mode which is querying all instruments on the bus.  And it's primary use is in automated test systems

You want to disable talk-only mode for normal interactive control of an instrument using GPIB

Content by Scott
Typos by Siri

On Feb 23, 2017, at 1:37 AM, James Peroulas james@peroulas.com wrote:

Does anyone have experience using older HP equipment with the NI
GPIB-ENET/1000 GPIB to ethernet box? Is this even possible with such an old
piece of equipment? I'm very familiar with using SCPI commands over telnet
but not so familiar with HPIB.

So far I've installed the NI 488.2 software which has successfully found
both the GPIB-ENET/1000 box and the HP5328A connected to it. I've opened
the NI-488.2 communicator which should allow me to send commands to the
counter and read back results. However, anything I type in the "send
string" returns a timeout.

The manual for the 5328A shows example commands such as:
CMD"?U*","T"
But it's not clear to me what portion (all?) is the string that I should be
sending to the 5328a.

The 5328a is set to "Talk Only" mode via dip switch. I would expect that I
should be able to just 'read' the results, but 'read' also times out...

Thanks for any pointers,
James


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

'Talk Only' mode responds to serial poll mode which is querying all instruments on the bus. And it's primary use is in automated test systems You want to disable talk-only mode for normal interactive control of an instrument using GPIB Content by Scott Typos by Siri > On Feb 23, 2017, at 1:37 AM, James Peroulas <james@peroulas.com> wrote: > > Does anyone have experience using older HP equipment with the NI > GPIB-ENET/1000 GPIB to ethernet box? Is this even possible with such an old > piece of equipment? I'm very familiar with using SCPI commands over telnet > but not so familiar with HPIB. > > So far I've installed the NI 488.2 software which has successfully found > both the GPIB-ENET/1000 box and the HP5328A connected to it. I've opened > the NI-488.2 communicator which should allow me to send commands to the > counter and read back results. However, anything I type in the "send > string" returns a timeout. > > The manual for the 5328A shows example commands such as: > CMD"?U*","T" > But it's not clear to me what portion (all?) is the string that I should be > sending to the 5328a. > > The 5328a is set to "Talk Only" mode via dip switch. I would expect that I > should be able to just 'read' the results, but 'read' also times out... > > Thanks for any pointers, > James > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
SS
Scott Stobbe
Thu, Feb 23, 2017 4:49 PM

If your counter is showing up in the NI device tree, you should be pretty
close to getting it working.

I haven't used a 5328A, but I suspect the command you have shown above is
sample code for a basic program, the actual HPIB command may just be a 'T'.

On Thu, Feb 23, 2017 at 1:37 AM, James Peroulas james@peroulas.com wrote:

Does anyone have experience using older HP equipment with the NI
GPIB-ENET/1000 GPIB to ethernet box? Is this even possible with such an old
piece of equipment? I'm very familiar with using SCPI commands over telnet
but not so familiar with HPIB.

So far I've installed the NI 488.2 software which has successfully found
both the GPIB-ENET/1000 box and the HP5328A connected to it. I've opened
the NI-488.2 communicator which should allow me to send commands to the
counter and read back results. However, anything I type in the "send
string" returns a timeout.

The manual for the 5328A shows example commands such as:
CMD"?U*","T"
But it's not clear to me what portion (all?) is the string that I should be
sending to the 5328a.

The 5328a is set to "Talk Only" mode via dip switch. I would expect that I
should be able to just 'read' the results, but 'read' also times out...

Thanks for any pointers,
James


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

If your counter is showing up in the NI device tree, you should be pretty close to getting it working. I haven't used a 5328A, but I suspect the command you have shown above is sample code for a basic program, the actual HPIB command may just be a 'T'. On Thu, Feb 23, 2017 at 1:37 AM, James Peroulas <james@peroulas.com> wrote: > Does anyone have experience using older HP equipment with the NI > GPIB-ENET/1000 GPIB to ethernet box? Is this even possible with such an old > piece of equipment? I'm very familiar with using SCPI commands over telnet > but not so familiar with HPIB. > > So far I've installed the NI 488.2 software which has successfully found > both the GPIB-ENET/1000 box and the HP5328A connected to it. I've opened > the NI-488.2 communicator which should allow me to send commands to the > counter and read back results. However, anything I type in the "send > string" returns a timeout. > > The manual for the 5328A shows example commands such as: > CMD"?U*","T" > But it's not clear to me what portion (all?) is the string that I should be > sending to the 5328a. > > The 5328a is set to "Talk Only" mode via dip switch. I would expect that I > should be able to just 'read' the results, but 'read' also times out... > > Thanks for any pointers, > James > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > and follow the instructions there. >