Hi. Apologies for a long post.

I'm trying to read up on the "care and feeding of hydrogen masers". While
they are conceptually simple from a distance, there's quite a bit going on
in the quantum mechanics department when looked at up close. Somewhat
frustratingly, I am not mentally equipped to really grasp the finer (or
even coarser) points of that particular department. The topic of this post
is the concept of spin exchange, and it's relation to cavity (auto) tuning.
I've read papers on the subject, but I am having difficulties building a
"workable intuition", so I turn to the group.

Here's what I think I understand, and I respectfully ask for corrections if
I am way off base here..

Spin-exchange in a hydrogen maser happens when two atoms collide, and
exchange spin, as it were.. (Hazy on the details here..) The number of
spin-exchange collisions is directly proportional to the density of atoms
in the cavity. These collisions will happen, but is a problem in hydrogen
masers for two reasons: 1) it takes away energy from the cavity, resulting
in lower signal output power, which degrades stability, and, 2) more
significantly, it results in a frequency shift.

The frequency shift, as far as I can gather, is directly related to the
cavity resonant frequency - there is no way to stop spin exchange taking
place (apart from reducing the hydrogen density to a level where collisions
are rare, in which case the density will be too low for oscillation to take
place), but it is possible to reduce the impact the spin exchange has on
the output frequency.

While the resonant frequency obviously influences the output power of the
maser cavity, the "mistuning" of the cavity also increases the effect spin
exchange has. In other words, in a perfectly tuned cavity, spin exchange
does not result in a frequency shift. In a badly tuned cavity, increasing
or decreasing the hydrogen flux (thereby increasing or decreasing the
number of collisions taking place) results in a corresponding
increase/decrease of the output frequency. Since the cavity ages, and the
cavity resonant frequency follows that aging, the long term stability of
the maser is degraded unless the aging can be compensated for. Which is
what cavity auto-tuning is all about.

From my understanding, there are a few ways to implement cavity auto-tuning:

  1. From the above, it follows that a modulation of the hydrogen flux into a
    mis-tuned cavity will result in a frequency shift following the modulation
    frequency. Using a stable reference, this shift can be measured, and
    corrections can be made to the cavity varactor voltage. Once the output
    frequency no longer shifts in response to the changes in hydrogen flux, the
    cavity is correctly tuned.

  2. It is also possible to modulate the cavity varactor voltage. By
    measuring the output power of the cavity, an error signal can be obtained
    and used to correct the average varactor voltage. A square wave of i.e.
    100hz, centered on the approximate correct varactor voltate is put in the
    varactor, and cavity output power is measured. If the output power measured
    on the "low" of the square wave is lower than the signal measured when the
    "high", lower the offset by some mV, and vice versa. Suitable filtering
    would of course be required.

The idea is that this method should not result in appreciable degradation
of the short/medium term stability of the maser, because the frequency of
the atoms interacting with the electromagnetic field in the maser cavity
takes time to respond to the changes in the resonant frequency, but the
output power responds "instantly". (Hazy on those details as well..) By
modulating the cavity varactor voltage (much) faster that the time constant
of the maser cavity, the modulation is effectively filtered out.

I am very interested in this method, as it seems to me that it would be
easy (feasible) to retrofit this to older masers never equipped with cavity
auto tuning.

There is at least one more way, which involves injecting a signal into the
maser cavity through a second coupling loop. At least one vendor I know of
does this in their newest design. I do not understand even the basics of
this method.

Any insights and/or corrections of my understanding is most welcome.

Thanks,
Ole

Hi. Apologies for a long post. I'm trying to read up on the "care and feeding of hydrogen masers". While they are conceptually simple from a distance, there's quite a bit going on in the quantum mechanics department when looked at up close. Somewhat frustratingly, I am not mentally equipped to really grasp the finer (or even coarser) points of that particular department. The topic of this post is the concept of spin exchange, and it's relation to cavity (auto) tuning. I've read papers on the subject, but I am having difficulties building a "workable intuition", so I turn to the group. Here's what I think I understand, and I respectfully ask for corrections if I am way off base here.. Spin-exchange in a hydrogen maser happens when two atoms collide, and exchange spin, as it were.. (Hazy on the details here..) The number of spin-exchange collisions is directly proportional to the density of atoms in the cavity. These collisions *will* happen, but is a problem in hydrogen masers for two reasons: 1) it takes away energy from the cavity, resulting in lower signal output power, which degrades stability, and, 2) more significantly, it results in a frequency shift. The frequency shift, as far as I can gather, is directly related to the cavity resonant frequency - there is no way to *stop* spin exchange taking place (apart from reducing the hydrogen density to a level where collisions are rare, in which case the density will be too low for oscillation to take place), but it is possible to reduce the impact the spin exchange has on the output frequency. While the resonant frequency obviously influences the output power of the maser cavity, the "mistuning" of the cavity also increases the effect spin exchange has. In other words, in a perfectly tuned cavity, spin exchange does not result in a frequency shift. In a badly tuned cavity, increasing or decreasing the hydrogen flux (thereby increasing or decreasing the number of collisions taking place) results in a corresponding increase/decrease of the output frequency. Since the cavity ages, and the cavity resonant frequency follows that aging, the long term stability of the maser is degraded unless the aging can be compensated for. Which is what cavity auto-tuning is all about. >From my understanding, there are a few ways to implement cavity auto-tuning: 1. From the above, it follows that a modulation of the hydrogen flux into a mis-tuned cavity will result in a frequency shift following the modulation frequency. Using a stable reference, this shift can be measured, and corrections can be made to the cavity varactor voltage. Once the output frequency no longer shifts in response to the changes in hydrogen flux, the cavity is correctly tuned. 2. It is also possible to modulate the cavity varactor voltage. By measuring the output power of the cavity, an error signal can be obtained and used to correct the average varactor voltage. A square wave of i.e. 100hz, centered on the approximate correct varactor voltate is put in the varactor, and cavity output power is measured. If the output power measured on the "low" of the square wave is lower than the signal measured when the "high", lower the offset by some mV, and vice versa. Suitable filtering would of course be required. The idea is that this method should not result in appreciable degradation of the short/medium term stability of the maser, because the frequency of the atoms interacting with the electromagnetic field in the maser cavity takes time to respond to the changes in the resonant frequency, but the output power responds "instantly". (Hazy on those details as well..) By modulating the cavity varactor voltage (much) faster that the time constant of the maser cavity, the modulation is effectively filtered out. I am very interested in this method, as it seems to me that it would be easy (feasible) to retrofit this to older masers never equipped with cavity auto tuning. There is at least one more way, which involves injecting a signal into the maser cavity through a second coupling loop. At least one vendor I know of does this in their newest design. I do not understand even the basics of this method. Any insights and/or corrections of my understanding is most welcome. Thanks, Ole
PS
paul swed
Fri, Jun 17, 2016 9:35 PM

Ole,
Quite looking forward to the replies. Though no intention to own a maser
currently.
I thought there was a method of rejecting or reducing types of spin.
Therefore reducing the impact you mention. Essentially a state selector.
Pretty sure that thought will get corrected pretty quickly.
Regards
Paul
WB8TSL

On Fri, Jun 17, 2016 at 3:33 PM, Ole Petter Ronningen <opronningen@gmail.com

wrote:

Hi. Apologies for a long post.

I'm trying to read up on the "care and feeding of hydrogen masers". While
they are conceptually simple from a distance, there's quite a bit going on
in the quantum mechanics department when looked at up close. Somewhat
frustratingly, I am not mentally equipped to really grasp the finer (or
even coarser) points of that particular department. The topic of this post
is the concept of spin exchange, and it's relation to cavity (auto) tuning.
I've read papers on the subject, but I am having difficulties building a
"workable intuition", so I turn to the group.

Here's what I think I understand, and I respectfully ask for corrections if
I am way off base here..

Spin-exchange in a hydrogen maser happens when two atoms collide, and
exchange spin, as it were.. (Hazy on the details here..) The number of
spin-exchange collisions is directly proportional to the density of atoms
in the cavity. These collisions will happen, but is a problem in hydrogen
masers for two reasons: 1) it takes away energy from the cavity, resulting
in lower signal output power, which degrades stability, and, 2) more
significantly, it results in a frequency shift.

The frequency shift, as far as I can gather, is directly related to the
cavity resonant frequency - there is no way to stop spin exchange taking
place (apart from reducing the hydrogen density to a level where collisions
are rare, in which case the density will be too low for oscillation to take
place), but it is possible to reduce the impact the spin exchange has on
the output frequency.

While the resonant frequency obviously influences the output power of the
maser cavity, the "mistuning" of the cavity also increases the effect spin
exchange has. In other words, in a perfectly tuned cavity, spin exchange
does not result in a frequency shift. In a badly tuned cavity, increasing
or decreasing the hydrogen flux (thereby increasing or decreasing the
number of collisions taking place) results in a corresponding
increase/decrease of the output frequency. Since the cavity ages, and the
cavity resonant frequency follows that aging, the long term stability of
the maser is degraded unless the aging can be compensated for. Which is
what cavity auto-tuning is all about.

From my understanding, there are a few ways to implement cavity
auto-tuning:

  1. From the above, it follows that a modulation of the hydrogen flux into a
    mis-tuned cavity will result in a frequency shift following the modulation
    frequency. Using a stable reference, this shift can be measured, and
    corrections can be made to the cavity varactor voltage. Once the output
    frequency no longer shifts in response to the changes in hydrogen flux, the
    cavity is correctly tuned.

  2. It is also possible to modulate the cavity varactor voltage. By
    measuring the output power of the cavity, an error signal can be obtained
    and used to correct the average varactor voltage. A square wave of i.e.
    100hz, centered on the approximate correct varactor voltate is put in the
    varactor, and cavity output power is measured. If the output power measured
    on the "low" of the square wave is lower than the signal measured when the
    "high", lower the offset by some mV, and vice versa. Suitable filtering
    would of course be required.

The idea is that this method should not result in appreciable degradation
of the short/medium term stability of the maser, because the frequency of
the atoms interacting with the electromagnetic field in the maser cavity
takes time to respond to the changes in the resonant frequency, but the
output power responds "instantly". (Hazy on those details as well..) By
modulating the cavity varactor voltage (much) faster that the time constant
of the maser cavity, the modulation is effectively filtered out.

I am very interested in this method, as it seems to me that it would be
easy (feasible) to retrofit this to older masers never equipped with cavity
auto tuning.

There is at least one more way, which involves injecting a signal into the
maser cavity through a second coupling loop. At least one vendor I know of
does this in their newest design. I do not understand even the basics of
this method.

Any insights and/or corrections of my understanding is most welcome.

Thanks,
Ole


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Ole, Quite looking forward to the replies. Though no intention to own a maser currently. I thought there was a method of rejecting or reducing types of spin. Therefore reducing the impact you mention. Essentially a state selector. Pretty sure that thought will get corrected pretty quickly. Regards Paul WB8TSL On Fri, Jun 17, 2016 at 3:33 PM, Ole Petter Ronningen <opronningen@gmail.com > wrote: > Hi. Apologies for a long post. > > I'm trying to read up on the "care and feeding of hydrogen masers". While > they are conceptually simple from a distance, there's quite a bit going on > in the quantum mechanics department when looked at up close. Somewhat > frustratingly, I am not mentally equipped to really grasp the finer (or > even coarser) points of that particular department. The topic of this post > is the concept of spin exchange, and it's relation to cavity (auto) tuning. > I've read papers on the subject, but I am having difficulties building a > "workable intuition", so I turn to the group. > > Here's what I think I understand, and I respectfully ask for corrections if > I am way off base here.. > > Spin-exchange in a hydrogen maser happens when two atoms collide, and > exchange spin, as it were.. (Hazy on the details here..) The number of > spin-exchange collisions is directly proportional to the density of atoms > in the cavity. These collisions *will* happen, but is a problem in hydrogen > masers for two reasons: 1) it takes away energy from the cavity, resulting > in lower signal output power, which degrades stability, and, 2) more > significantly, it results in a frequency shift. > > The frequency shift, as far as I can gather, is directly related to the > cavity resonant frequency - there is no way to *stop* spin exchange taking > place (apart from reducing the hydrogen density to a level where collisions > are rare, in which case the density will be too low for oscillation to take > place), but it is possible to reduce the impact the spin exchange has on > the output frequency. > > While the resonant frequency obviously influences the output power of the > maser cavity, the "mistuning" of the cavity also increases the effect spin > exchange has. In other words, in a perfectly tuned cavity, spin exchange > does not result in a frequency shift. In a badly tuned cavity, increasing > or decreasing the hydrogen flux (thereby increasing or decreasing the > number of collisions taking place) results in a corresponding > increase/decrease of the output frequency. Since the cavity ages, and the > cavity resonant frequency follows that aging, the long term stability of > the maser is degraded unless the aging can be compensated for. Which is > what cavity auto-tuning is all about. > > From my understanding, there are a few ways to implement cavity > auto-tuning: > 1. From the above, it follows that a modulation of the hydrogen flux into a > mis-tuned cavity will result in a frequency shift following the modulation > frequency. Using a stable reference, this shift can be measured, and > corrections can be made to the cavity varactor voltage. Once the output > frequency no longer shifts in response to the changes in hydrogen flux, the > cavity is correctly tuned. > > 2. It is also possible to modulate the cavity varactor voltage. By > measuring the output power of the cavity, an error signal can be obtained > and used to correct the average varactor voltage. A square wave of i.e. > 100hz, centered on the approximate correct varactor voltate is put in the > varactor, and cavity output power is measured. If the output power measured > on the "low" of the square wave is lower than the signal measured when the > "high", lower the offset by some mV, and vice versa. Suitable filtering > would of course be required. > > The idea is that this method should not result in appreciable degradation > of the short/medium term stability of the maser, because the frequency of > the atoms interacting with the electromagnetic field in the maser cavity > takes time to respond to the changes in the resonant frequency, but the > output power responds "instantly". (Hazy on those details as well..) By > modulating the cavity varactor voltage (much) faster that the time constant > of the maser cavity, the modulation is effectively filtered out. > > I am very interested in this method, as it seems to me that it would be > easy (feasible) to retrofit this to older masers never equipped with cavity > auto tuning. > > There is at least one more way, which involves injecting a signal into the > maser cavity through a second coupling loop. At least one vendor I know of > does this in their newest design. I do not understand even the basics of > this method. > > Any insights and/or corrections of my understanding is most welcome. > > Thanks, > Ole > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
OP
Ole Petter Ronningen
Fri, Jun 17, 2016 10:18 PM

There is indeed a quad- or hexa-pole state selection magnet between the
dissociator and the cavity, required to select the atoms of the correct
state to allow a population inversion resulting in oscillation.

My understanding is that the spin-exchange of interest takes place between
atoms already bouncing around inside the cavity - having passed that
particular obstacle.

And while on that subject, if anyone could elaborate on the difference
between the two (quad or hexa-pole), that would be interesting. I have not
found a clear explanation for why one might choose one over the other.

Ole

On Fri, Jun 17, 2016 at 11:35 PM, paul swed paulswedb@gmail.com wrote:

Ole,
Quite looking forward to the replies. Though no intention to own a maser
currently.
I thought there was a method of rejecting or reducing types of spin.
Therefore reducing the impact you mention. Essentially a state selector.
Pretty sure that thought will get corrected pretty quickly.
Regards
Paul
WB8TSL

On Fri, Jun 17, 2016 at 3:33 PM, Ole Petter Ronningen <
opronningen@gmail.com

wrote:

Hi. Apologies for a long post.

I'm trying to read up on the "care and feeding of hydrogen masers". While
they are conceptually simple from a distance, there's quite a bit going

on

in the quantum mechanics department when looked at up close. Somewhat
frustratingly, I am not mentally equipped to really grasp the finer (or
even coarser) points of that particular department. The topic of this

post

is the concept of spin exchange, and it's relation to cavity (auto)

tuning.

I've read papers on the subject, but I am having difficulties building a
"workable intuition", so I turn to the group.

Here's what I think I understand, and I respectfully ask for corrections

if

I am way off base here..

Spin-exchange in a hydrogen maser happens when two atoms collide, and
exchange spin, as it were.. (Hazy on the details here..) The number of
spin-exchange collisions is directly proportional to the density of atoms
in the cavity. These collisions will happen, but is a problem in

hydrogen

masers for two reasons: 1) it takes away energy from the cavity,

resulting

in lower signal output power, which degrades stability, and, 2) more
significantly, it results in a frequency shift.

The frequency shift, as far as I can gather, is directly related to the
cavity resonant frequency - there is no way to stop spin exchange

taking

place (apart from reducing the hydrogen density to a level where

collisions

are rare, in which case the density will be too low for oscillation to

take

place), but it is possible to reduce the impact the spin exchange has on
the output frequency.

While the resonant frequency obviously influences the output power of the
maser cavity, the "mistuning" of the cavity also increases the effect

spin

exchange has. In other words, in a perfectly tuned cavity, spin exchange
does not result in a frequency shift. In a badly tuned cavity, increasing
or decreasing the hydrogen flux (thereby increasing or decreasing the
number of collisions taking place) results in a corresponding
increase/decrease of the output frequency. Since the cavity ages, and the
cavity resonant frequency follows that aging, the long term stability of
the maser is degraded unless the aging can be compensated for. Which is
what cavity auto-tuning is all about.

From my understanding, there are a few ways to implement cavity
auto-tuning:

  1. From the above, it follows that a modulation of the hydrogen flux

into a

mis-tuned cavity will result in a frequency shift following the

modulation

frequency. Using a stable reference, this shift can be measured, and
corrections can be made to the cavity varactor voltage. Once the output
frequency no longer shifts in response to the changes in hydrogen flux,

the

cavity is correctly tuned.

  1. It is also possible to modulate the cavity varactor voltage. By
    measuring the output power of the cavity, an error signal can be obtained
    and used to correct the average varactor voltage. A square wave of i.e.
    100hz, centered on the approximate correct varactor voltate is put in the
    varactor, and cavity output power is measured. If the output power

measured

on the "low" of the square wave is lower than the signal measured when

the

"high", lower the offset by some mV, and vice versa. Suitable filtering
would of course be required.

The idea is that this method should not result in appreciable degradation
of the short/medium term stability of the maser, because the frequency of
the atoms interacting with the electromagnetic field in the maser cavity
takes time to respond to the changes in the resonant frequency, but the
output power responds "instantly". (Hazy on those details as well..) By
modulating the cavity varactor voltage (much) faster that the time

constant

of the maser cavity, the modulation is effectively filtered out.

I am very interested in this method, as it seems to me that it would be
easy (feasible) to retrofit this to older masers never equipped with

cavity

auto tuning.

There is at least one more way, which involves injecting a signal into

the

maser cavity through a second coupling loop. At least one vendor I know

of

does this in their newest design. I do not understand even the basics of
this method.

Any insights and/or corrections of my understanding is most welcome.

Thanks,
Ole


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

There is indeed a quad- or hexa-pole state selection magnet between the dissociator and the cavity, required to select the atoms of the correct state to allow a population inversion resulting in oscillation. My understanding is that the spin-exchange of interest takes place between atoms already bouncing around inside the cavity - having passed that particular obstacle. And while on that subject, if anyone could elaborate on the difference between the two (quad or hexa-pole), that would be interesting. I have not found a clear explanation for why one might choose one over the other. Ole On Fri, Jun 17, 2016 at 11:35 PM, paul swed <paulswedb@gmail.com> wrote: > Ole, > Quite looking forward to the replies. Though no intention to own a maser > currently. > I thought there was a method of rejecting or reducing types of spin. > Therefore reducing the impact you mention. Essentially a state selector. > Pretty sure that thought will get corrected pretty quickly. > Regards > Paul > WB8TSL > > On Fri, Jun 17, 2016 at 3:33 PM, Ole Petter Ronningen < > opronningen@gmail.com > > wrote: > > > Hi. Apologies for a long post. > > > > I'm trying to read up on the "care and feeding of hydrogen masers". While > > they are conceptually simple from a distance, there's quite a bit going > on > > in the quantum mechanics department when looked at up close. Somewhat > > frustratingly, I am not mentally equipped to really grasp the finer (or > > even coarser) points of that particular department. The topic of this > post > > is the concept of spin exchange, and it's relation to cavity (auto) > tuning. > > I've read papers on the subject, but I am having difficulties building a > > "workable intuition", so I turn to the group. > > > > Here's what I think I understand, and I respectfully ask for corrections > if > > I am way off base here.. > > > > Spin-exchange in a hydrogen maser happens when two atoms collide, and > > exchange spin, as it were.. (Hazy on the details here..) The number of > > spin-exchange collisions is directly proportional to the density of atoms > > in the cavity. These collisions *will* happen, but is a problem in > hydrogen > > masers for two reasons: 1) it takes away energy from the cavity, > resulting > > in lower signal output power, which degrades stability, and, 2) more > > significantly, it results in a frequency shift. > > > > The frequency shift, as far as I can gather, is directly related to the > > cavity resonant frequency - there is no way to *stop* spin exchange > taking > > place (apart from reducing the hydrogen density to a level where > collisions > > are rare, in which case the density will be too low for oscillation to > take > > place), but it is possible to reduce the impact the spin exchange has on > > the output frequency. > > > > While the resonant frequency obviously influences the output power of the > > maser cavity, the "mistuning" of the cavity also increases the effect > spin > > exchange has. In other words, in a perfectly tuned cavity, spin exchange > > does not result in a frequency shift. In a badly tuned cavity, increasing > > or decreasing the hydrogen flux (thereby increasing or decreasing the > > number of collisions taking place) results in a corresponding > > increase/decrease of the output frequency. Since the cavity ages, and the > > cavity resonant frequency follows that aging, the long term stability of > > the maser is degraded unless the aging can be compensated for. Which is > > what cavity auto-tuning is all about. > > > > From my understanding, there are a few ways to implement cavity > > auto-tuning: > > 1. From the above, it follows that a modulation of the hydrogen flux > into a > > mis-tuned cavity will result in a frequency shift following the > modulation > > frequency. Using a stable reference, this shift can be measured, and > > corrections can be made to the cavity varactor voltage. Once the output > > frequency no longer shifts in response to the changes in hydrogen flux, > the > > cavity is correctly tuned. > > > > 2. It is also possible to modulate the cavity varactor voltage. By > > measuring the output power of the cavity, an error signal can be obtained > > and used to correct the average varactor voltage. A square wave of i.e. > > 100hz, centered on the approximate correct varactor voltate is put in the > > varactor, and cavity output power is measured. If the output power > measured > > on the "low" of the square wave is lower than the signal measured when > the > > "high", lower the offset by some mV, and vice versa. Suitable filtering > > would of course be required. > > > > The idea is that this method should not result in appreciable degradation > > of the short/medium term stability of the maser, because the frequency of > > the atoms interacting with the electromagnetic field in the maser cavity > > takes time to respond to the changes in the resonant frequency, but the > > output power responds "instantly". (Hazy on those details as well..) By > > modulating the cavity varactor voltage (much) faster that the time > constant > > of the maser cavity, the modulation is effectively filtered out. > > > > I am very interested in this method, as it seems to me that it would be > > easy (feasible) to retrofit this to older masers never equipped with > cavity > > auto tuning. > > > > There is at least one more way, which involves injecting a signal into > the > > maser cavity through a second coupling loop. At least one vendor I know > of > > does this in their newest design. I do not understand even the basics of > > this method. > > > > Any insights and/or corrections of my understanding is most welcome. > > > > Thanks, > > Ole > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
PS
paul swed
Sat, Jun 18, 2016 12:43 AM

Ole I can only point you toward cesium references I remember reading about
the magnet configuration and it did matter. Sorry it was a while ago.
Others may know the reference. It was the HP 5060  as I barely recall.

On Fri, Jun 17, 2016 at 6:18 PM, Ole Petter Ronningen <opronningen@gmail.com

wrote:

There is indeed a quad- or hexa-pole state selection magnet between the
dissociator and the cavity, required to select the atoms of the correct
state to allow a population inversion resulting in oscillation.

My understanding is that the spin-exchange of interest takes place between
atoms already bouncing around inside the cavity - having passed that
particular obstacle.

And while on that subject, if anyone could elaborate on the difference
between the two (quad or hexa-pole), that would be interesting. I have not
found a clear explanation for why one might choose one over the other.

Ole

On Fri, Jun 17, 2016 at 11:35 PM, paul swed paulswedb@gmail.com wrote:

Ole,
Quite looking forward to the replies. Though no intention to own a maser
currently.
I thought there was a method of rejecting or reducing types of spin.
Therefore reducing the impact you mention. Essentially a state selector.
Pretty sure that thought will get corrected pretty quickly.
Regards
Paul
WB8TSL

On Fri, Jun 17, 2016 at 3:33 PM, Ole Petter Ronningen <
opronningen@gmail.com

wrote:

Hi. Apologies for a long post.

I'm trying to read up on the "care and feeding of hydrogen masers".

While

they are conceptually simple from a distance, there's quite a bit going

on

in the quantum mechanics department when looked at up close. Somewhat
frustratingly, I am not mentally equipped to really grasp the finer (or
even coarser) points of that particular department. The topic of this

post

is the concept of spin exchange, and it's relation to cavity (auto)

tuning.

I've read papers on the subject, but I am having difficulties building

a

"workable intuition", so I turn to the group.

Here's what I think I understand, and I respectfully ask for

corrections

if

I am way off base here..

Spin-exchange in a hydrogen maser happens when two atoms collide, and
exchange spin, as it were.. (Hazy on the details here..) The number of
spin-exchange collisions is directly proportional to the density of

atoms

in the cavity. These collisions will happen, but is a problem in

hydrogen

masers for two reasons: 1) it takes away energy from the cavity,

resulting

in lower signal output power, which degrades stability, and, 2) more
significantly, it results in a frequency shift.

The frequency shift, as far as I can gather, is directly related to the
cavity resonant frequency - there is no way to stop spin exchange

taking

place (apart from reducing the hydrogen density to a level where

collisions

are rare, in which case the density will be too low for oscillation to

take

place), but it is possible to reduce the impact the spin exchange has

on

the output frequency.

While the resonant frequency obviously influences the output power of

the

maser cavity, the "mistuning" of the cavity also increases the effect

spin

exchange has. In other words, in a perfectly tuned cavity, spin

exchange

does not result in a frequency shift. In a badly tuned cavity,

increasing

or decreasing the hydrogen flux (thereby increasing or decreasing the
number of collisions taking place) results in a corresponding
increase/decrease of the output frequency. Since the cavity ages, and

the

cavity resonant frequency follows that aging, the long term stability

of

the maser is degraded unless the aging can be compensated for. Which is
what cavity auto-tuning is all about.

From my understanding, there are a few ways to implement cavity
auto-tuning:

  1. From the above, it follows that a modulation of the hydrogen flux

into a

mis-tuned cavity will result in a frequency shift following the

modulation

frequency. Using a stable reference, this shift can be measured, and
corrections can be made to the cavity varactor voltage. Once the output
frequency no longer shifts in response to the changes in hydrogen flux,

the

cavity is correctly tuned.

  1. It is also possible to modulate the cavity varactor voltage. By
    measuring the output power of the cavity, an error signal can be

obtained

and used to correct the average varactor voltage. A square wave of i.e.
100hz, centered on the approximate correct varactor voltate is put in

the

varactor, and cavity output power is measured. If the output power

measured

on the "low" of the square wave is lower than the signal measured when

the

"high", lower the offset by some mV, and vice versa. Suitable filtering
would of course be required.

The idea is that this method should not result in appreciable

degradation

of the short/medium term stability of the maser, because the frequency

of

the atoms interacting with the electromagnetic field in the maser

cavity

takes time to respond to the changes in the resonant frequency, but the
output power responds "instantly". (Hazy on those details as well..) By
modulating the cavity varactor voltage (much) faster that the time

constant

of the maser cavity, the modulation is effectively filtered out.

I am very interested in this method, as it seems to me that it would be
easy (feasible) to retrofit this to older masers never equipped with

cavity

auto tuning.

There is at least one more way, which involves injecting a signal into

the

maser cavity through a second coupling loop. At least one vendor I know

of

does this in their newest design. I do not understand even the basics

of

this method.

Any insights and/or corrections of my understanding is most welcome.

Thanks,
Ole


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Ole I can only point you toward cesium references I remember reading about the magnet configuration and it did matter. Sorry it was a while ago. Others may know the reference. It was the HP 5060 as I barely recall. On Fri, Jun 17, 2016 at 6:18 PM, Ole Petter Ronningen <opronningen@gmail.com > wrote: > There is indeed a quad- or hexa-pole state selection magnet between the > dissociator and the cavity, required to select the atoms of the correct > state to allow a population inversion resulting in oscillation. > > My understanding is that the spin-exchange of interest takes place between > atoms already bouncing around inside the cavity - having passed that > particular obstacle. > > And while on that subject, if anyone could elaborate on the difference > between the two (quad or hexa-pole), that would be interesting. I have not > found a clear explanation for why one might choose one over the other. > > Ole > > On Fri, Jun 17, 2016 at 11:35 PM, paul swed <paulswedb@gmail.com> wrote: > > > Ole, > > Quite looking forward to the replies. Though no intention to own a maser > > currently. > > I thought there was a method of rejecting or reducing types of spin. > > Therefore reducing the impact you mention. Essentially a state selector. > > Pretty sure that thought will get corrected pretty quickly. > > Regards > > Paul > > WB8TSL > > > > On Fri, Jun 17, 2016 at 3:33 PM, Ole Petter Ronningen < > > opronningen@gmail.com > > > wrote: > > > > > Hi. Apologies for a long post. > > > > > > I'm trying to read up on the "care and feeding of hydrogen masers". > While > > > they are conceptually simple from a distance, there's quite a bit going > > on > > > in the quantum mechanics department when looked at up close. Somewhat > > > frustratingly, I am not mentally equipped to really grasp the finer (or > > > even coarser) points of that particular department. The topic of this > > post > > > is the concept of spin exchange, and it's relation to cavity (auto) > > tuning. > > > I've read papers on the subject, but I am having difficulties building > a > > > "workable intuition", so I turn to the group. > > > > > > Here's what I think I understand, and I respectfully ask for > corrections > > if > > > I am way off base here.. > > > > > > Spin-exchange in a hydrogen maser happens when two atoms collide, and > > > exchange spin, as it were.. (Hazy on the details here..) The number of > > > spin-exchange collisions is directly proportional to the density of > atoms > > > in the cavity. These collisions *will* happen, but is a problem in > > hydrogen > > > masers for two reasons: 1) it takes away energy from the cavity, > > resulting > > > in lower signal output power, which degrades stability, and, 2) more > > > significantly, it results in a frequency shift. > > > > > > The frequency shift, as far as I can gather, is directly related to the > > > cavity resonant frequency - there is no way to *stop* spin exchange > > taking > > > place (apart from reducing the hydrogen density to a level where > > collisions > > > are rare, in which case the density will be too low for oscillation to > > take > > > place), but it is possible to reduce the impact the spin exchange has > on > > > the output frequency. > > > > > > While the resonant frequency obviously influences the output power of > the > > > maser cavity, the "mistuning" of the cavity also increases the effect > > spin > > > exchange has. In other words, in a perfectly tuned cavity, spin > exchange > > > does not result in a frequency shift. In a badly tuned cavity, > increasing > > > or decreasing the hydrogen flux (thereby increasing or decreasing the > > > number of collisions taking place) results in a corresponding > > > increase/decrease of the output frequency. Since the cavity ages, and > the > > > cavity resonant frequency follows that aging, the long term stability > of > > > the maser is degraded unless the aging can be compensated for. Which is > > > what cavity auto-tuning is all about. > > > > > > From my understanding, there are a few ways to implement cavity > > > auto-tuning: > > > 1. From the above, it follows that a modulation of the hydrogen flux > > into a > > > mis-tuned cavity will result in a frequency shift following the > > modulation > > > frequency. Using a stable reference, this shift can be measured, and > > > corrections can be made to the cavity varactor voltage. Once the output > > > frequency no longer shifts in response to the changes in hydrogen flux, > > the > > > cavity is correctly tuned. > > > > > > 2. It is also possible to modulate the cavity varactor voltage. By > > > measuring the output power of the cavity, an error signal can be > obtained > > > and used to correct the average varactor voltage. A square wave of i.e. > > > 100hz, centered on the approximate correct varactor voltate is put in > the > > > varactor, and cavity output power is measured. If the output power > > measured > > > on the "low" of the square wave is lower than the signal measured when > > the > > > "high", lower the offset by some mV, and vice versa. Suitable filtering > > > would of course be required. > > > > > > The idea is that this method should not result in appreciable > degradation > > > of the short/medium term stability of the maser, because the frequency > of > > > the atoms interacting with the electromagnetic field in the maser > cavity > > > takes time to respond to the changes in the resonant frequency, but the > > > output power responds "instantly". (Hazy on those details as well..) By > > > modulating the cavity varactor voltage (much) faster that the time > > constant > > > of the maser cavity, the modulation is effectively filtered out. > > > > > > I am very interested in this method, as it seems to me that it would be > > > easy (feasible) to retrofit this to older masers never equipped with > > cavity > > > auto tuning. > > > > > > There is at least one more way, which involves injecting a signal into > > the > > > maser cavity through a second coupling loop. At least one vendor I know > > of > > > does this in their newest design. I do not understand even the basics > of > > > this method. > > > > > > Any insights and/or corrections of my understanding is most welcome. > > > > > > Thanks, > > > Ole > > > _______________________________________________ > > > time-nuts mailing list -- time-nuts@febo.com > > > To unsubscribe, go to > > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > > and follow the instructions there. > > > > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to > > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
BC
Bob Camp
Sat, Jun 18, 2016 7:02 PM

Hi

Far more than you ever wanted to know about state selectors:

http://ntrs.nasa.gov/archive/nasa/casi.ntrs.nasa.gov/19820012652.pdf

Bob

On Jun 17, 2016, at 3:33 PM, Ole Petter Ronningen opronningen@gmail.com wrote:

Hi. Apologies for a long post.

I'm trying to read up on the "care and feeding of hydrogen masers". While
they are conceptually simple from a distance, there's quite a bit going on
in the quantum mechanics department when looked at up close. Somewhat
frustratingly, I am not mentally equipped to really grasp the finer (or
even coarser) points of that particular department. The topic of this post
is the concept of spin exchange, and it's relation to cavity (auto) tuning.
I've read papers on the subject, but I am having difficulties building a
"workable intuition", so I turn to the group.

Here's what I think I understand, and I respectfully ask for corrections if
I am way off base here..

Spin-exchange in a hydrogen maser happens when two atoms collide, and
exchange spin, as it were.. (Hazy on the details here..) The number of
spin-exchange collisions is directly proportional to the density of atoms
in the cavity. These collisions will happen, but is a problem in hydrogen
masers for two reasons: 1) it takes away energy from the cavity, resulting
in lower signal output power, which degrades stability, and, 2) more
significantly, it results in a frequency shift.

The frequency shift, as far as I can gather, is directly related to the
cavity resonant frequency - there is no way to stop spin exchange taking
place (apart from reducing the hydrogen density to a level where collisions
are rare, in which case the density will be too low for oscillation to take
place), but it is possible to reduce the impact the spin exchange has on
the output frequency.

While the resonant frequency obviously influences the output power of the
maser cavity, the "mistuning" of the cavity also increases the effect spin
exchange has. In other words, in a perfectly tuned cavity, spin exchange
does not result in a frequency shift. In a badly tuned cavity, increasing
or decreasing the hydrogen flux (thereby increasing or decreasing the
number of collisions taking place) results in a corresponding
increase/decrease of the output frequency. Since the cavity ages, and the
cavity resonant frequency follows that aging, the long term stability of
the maser is degraded unless the aging can be compensated for. Which is
what cavity auto-tuning is all about.

From my understanding, there are a few ways to implement cavity auto-tuning:

  1. From the above, it follows that a modulation of the hydrogen flux into a
    mis-tuned cavity will result in a frequency shift following the modulation
    frequency. Using a stable reference, this shift can be measured, and
    corrections can be made to the cavity varactor voltage. Once the output
    frequency no longer shifts in response to the changes in hydrogen flux, the
    cavity is correctly tuned.

  2. It is also possible to modulate the cavity varactor voltage. By
    measuring the output power of the cavity, an error signal can be obtained
    and used to correct the average varactor voltage. A square wave of i.e.
    100hz, centered on the approximate correct varactor voltate is put in the
    varactor, and cavity output power is measured. If the output power measured
    on the "low" of the square wave is lower than the signal measured when the
    "high", lower the offset by some mV, and vice versa. Suitable filtering
    would of course be required.

The idea is that this method should not result in appreciable degradation
of the short/medium term stability of the maser, because the frequency of
the atoms interacting with the electromagnetic field in the maser cavity
takes time to respond to the changes in the resonant frequency, but the
output power responds "instantly". (Hazy on those details as well..) By
modulating the cavity varactor voltage (much) faster that the time constant
of the maser cavity, the modulation is effectively filtered out.

I am very interested in this method, as it seems to me that it would be
easy (feasible) to retrofit this to older masers never equipped with cavity
auto tuning.

There is at least one more way, which involves injecting a signal into the
maser cavity through a second coupling loop. At least one vendor I know of
does this in their newest design. I do not understand even the basics of
this method.

Any insights and/or corrections of my understanding is most welcome.

Thanks,
Ole


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Far more than you *ever* wanted to know about state selectors: http://ntrs.nasa.gov/archive/nasa/casi.ntrs.nasa.gov/19820012652.pdf Bob > On Jun 17, 2016, at 3:33 PM, Ole Petter Ronningen <opronningen@gmail.com> wrote: > > Hi. Apologies for a long post. > > I'm trying to read up on the "care and feeding of hydrogen masers". While > they are conceptually simple from a distance, there's quite a bit going on > in the quantum mechanics department when looked at up close. Somewhat > frustratingly, I am not mentally equipped to really grasp the finer (or > even coarser) points of that particular department. The topic of this post > is the concept of spin exchange, and it's relation to cavity (auto) tuning. > I've read papers on the subject, but I am having difficulties building a > "workable intuition", so I turn to the group. > > Here's what I think I understand, and I respectfully ask for corrections if > I am way off base here.. > > Spin-exchange in a hydrogen maser happens when two atoms collide, and > exchange spin, as it were.. (Hazy on the details here..) The number of > spin-exchange collisions is directly proportional to the density of atoms > in the cavity. These collisions *will* happen, but is a problem in hydrogen > masers for two reasons: 1) it takes away energy from the cavity, resulting > in lower signal output power, which degrades stability, and, 2) more > significantly, it results in a frequency shift. > > The frequency shift, as far as I can gather, is directly related to the > cavity resonant frequency - there is no way to *stop* spin exchange taking > place (apart from reducing the hydrogen density to a level where collisions > are rare, in which case the density will be too low for oscillation to take > place), but it is possible to reduce the impact the spin exchange has on > the output frequency. > > While the resonant frequency obviously influences the output power of the > maser cavity, the "mistuning" of the cavity also increases the effect spin > exchange has. In other words, in a perfectly tuned cavity, spin exchange > does not result in a frequency shift. In a badly tuned cavity, increasing > or decreasing the hydrogen flux (thereby increasing or decreasing the > number of collisions taking place) results in a corresponding > increase/decrease of the output frequency. Since the cavity ages, and the > cavity resonant frequency follows that aging, the long term stability of > the maser is degraded unless the aging can be compensated for. Which is > what cavity auto-tuning is all about. > > From my understanding, there are a few ways to implement cavity auto-tuning: > 1. From the above, it follows that a modulation of the hydrogen flux into a > mis-tuned cavity will result in a frequency shift following the modulation > frequency. Using a stable reference, this shift can be measured, and > corrections can be made to the cavity varactor voltage. Once the output > frequency no longer shifts in response to the changes in hydrogen flux, the > cavity is correctly tuned. > > 2. It is also possible to modulate the cavity varactor voltage. By > measuring the output power of the cavity, an error signal can be obtained > and used to correct the average varactor voltage. A square wave of i.e. > 100hz, centered on the approximate correct varactor voltate is put in the > varactor, and cavity output power is measured. If the output power measured > on the "low" of the square wave is lower than the signal measured when the > "high", lower the offset by some mV, and vice versa. Suitable filtering > would of course be required. > > The idea is that this method should not result in appreciable degradation > of the short/medium term stability of the maser, because the frequency of > the atoms interacting with the electromagnetic field in the maser cavity > takes time to respond to the changes in the resonant frequency, but the > output power responds "instantly". (Hazy on those details as well..) By > modulating the cavity varactor voltage (much) faster that the time constant > of the maser cavity, the modulation is effectively filtered out. > > I am very interested in this method, as it seems to me that it would be > easy (feasible) to retrofit this to older masers never equipped with cavity > auto tuning. > > There is at least one more way, which involves injecting a signal into the > maser cavity through a second coupling loop. At least one vendor I know of > does this in their newest design. I do not understand even the basics of > this method. > > Any insights and/or corrections of my understanding is most welcome. > > Thanks, > Ole > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.