MS
Mark Sims
Fri, Aug 19, 2016 9:29 PM

Try driving your counter frequency reference input with the LPRO or comparing the LPRO with the GPSDO.  The LPRO is (most likely) much more accurate the counters' internal timebase.

Try driving your counter frequency reference input with the LPRO or comparing the LPRO with the GPSDO. The LPRO is (most likely) much more accurate the counters' internal timebase.
BK
Bob kb8tq
Sat, Aug 20, 2016 12:03 AM

Hi

Ok let's toss some numbers into the mix.

The counter time base one day after calibration is in the 0.5 to 1.5 ppb range.

The LPRO ten years after it left the factory is in the 0.5 to 1.5 ppb range.

The GPSDO when running properly should be in the 0.005 to 0.02 ppb range at one second.

One ppb at 10 MHz is 0.01 Hz. It also is a the resolution limit on a 5335 at 1 second.

Bob

On Aug 19, 2016, at 5:29 PM, Mark Sims holrum@hotmail.com wrote:

Try driving your counter frequency reference input with the LPRO or comparing the LPRO with the GPSDO.  The LPRO is (most likely) much more accurate the counters' internal timebase.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Ok let's toss some numbers into the mix. The counter time base one day after calibration is in the 0.5 to 1.5 ppb range. The LPRO ten years after it left the factory is in the 0.5 to 1.5 ppb range. The GPSDO when running properly should be in the 0.005 to 0.02 ppb range at one second. One ppb at 10 MHz is 0.01 Hz. It also is a the resolution limit on a 5335 at 1 second. Bob > On Aug 19, 2016, at 5:29 PM, Mark Sims <holrum@hotmail.com> wrote: > > Try driving your counter frequency reference input with the LPRO or comparing the LPRO with the GPSDO. The LPRO is (most likely) much more accurate the counters' internal timebase. > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
B_
Bryan _
Sat, Aug 20, 2016 9:43 AM

Bob:
With a GPSDO as a general rule is there much variation of the 0.005 to 0.02 ppb range. Would there me much if only 2 satellites were visible but the unit was no in holdover vs say 6 satellites. I assume there is probably a lot of variables such as quality of GPSDO and OCXO, multi path interference, antenna location, etc, but worse case what would a user expect assuming the unit isn't in holdover.

-=Bryan=-

From: kb8tq@n1k.org
Date: Fri, 19 Aug 2016 20:03:59 -0400
To: time-nuts@febo.com
Subject: Re: [time-nuts] Expected 10 MHz offset from a GPSDO?

Hi

Ok let's toss some numbers into the mix.

The counter time base one day after calibration is in the 0.5 to 1.5 ppb range.

The LPRO ten years after it left the factory is in the 0.5 to 1.5 ppb range.

The GPSDO when running properly should be in the 0.005 to 0.02 ppb range at one second.

One ppb at 10 MHz is 0.01 Hz. It also is a the resolution limit on a 5335 at 1 second.

Bob

On Aug 19, 2016, at 5:29 PM, Mark Sims holrum@hotmail.com wrote:

Try driving your counter frequency reference input with the LPRO or comparing the LPRO with the GPSDO.  The LPRO is (most likely) much more accurate the counters' internal timebase.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Bob: With a GPSDO as a general rule is there much variation of the 0.005 to 0.02 ppb range. Would there me much if only 2 satellites were visible but the unit was no in holdover vs say 6 satellites. I assume there is probably a lot of variables such as quality of GPSDO and OCXO, multi path interference, antenna location, etc, but worse case what would a user expect assuming the unit isn't in holdover. -=Bryan=- > From: kb8tq@n1k.org > Date: Fri, 19 Aug 2016 20:03:59 -0400 > To: time-nuts@febo.com > Subject: Re: [time-nuts] Expected 10 MHz offset from a GPSDO? > > Hi > > Ok let's toss some numbers into the mix. > > The counter time base one day after calibration is in the 0.5 to 1.5 ppb range. > > The LPRO ten years after it left the factory is in the 0.5 to 1.5 ppb range. > > The GPSDO when running properly should be in the 0.005 to 0.02 ppb range at one second. > > One ppb at 10 MHz is 0.01 Hz. It also is a the resolution limit on a 5335 at 1 second. > > Bob > > > > > On Aug 19, 2016, at 5:29 PM, Mark Sims <holrum@hotmail.com> wrote: > > > > Try driving your counter frequency reference input with the LPRO or comparing the LPRO with the GPSDO. The LPRO is (most likely) much more accurate the counters' internal timebase. > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > > and follow the instructions there. > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
CS
Charles Steinmetz
Sat, Aug 20, 2016 11:38 AM

Bryan wrote:

With a GPSDO as a general rule is there much variation of the 0.005 to 0.02 ppb range. Would there me much if only 2 satellites were visible but the unit was no in holdover vs say 6 satellites. I assume there is probably a lot of variables such as quality of GPSDO and OCXO, multi path interference, antenna location, etc, but worse case what would a user expect assuming the unit isn't in holdover.

There are lots of ADEV plots of GPSDOs on the web, if you search for
them.  Tom has the attached plot posted on his web site -- it includes
one decent (but not stellar) GPSDO, the HP Z3801.  That is about what
you can expect without heroic tuning efforts (on a GPSDO that supports
user fiddling, such as the Trimble Thunderbolt).

Since many counters gate for one second (or gate faster, then average),
the ADEV at tau=1 second is a reasonable estimate of the real-world
performance of a good counter driven by a GPSDO.  Note that, since the
ADEV at 1 second depends largely on the OCXO, the heroic tuning efforts
mentioned above won't change the ADEV at tau=1 second hardly at all.
That is attributable strictly to the OCXO (unless the GPSDO is very
poorly designed).  Building a GPSDO using the best OCXO available can
improve things.  For example, look at the "BVA" plot -- at 1 second, its
ADEV is nearly a full decade better than the Z3801, better even than the
hydrogen masers.  (While I referred to the 3801 above as "decent but not
stellar," that is based on its performance at tau from 100-100k seconds.
Its OCXO -- and, thus, its ADEV -- is better than most others at 1
second.)

Best regards,

Charles

Bryan wrote: > With a GPSDO as a general rule is there much variation of the 0.005 to 0.02 ppb range. Would there me much if only 2 satellites were visible but the unit was no in holdover vs say 6 satellites. I assume there is probably a lot of variables such as quality of GPSDO and OCXO, multi path interference, antenna location, etc, but worse case what would a user expect assuming the unit isn't in holdover. There are lots of ADEV plots of GPSDOs on the web, if you search for them. Tom has the attached plot posted on his web site -- it includes one decent (but not stellar) GPSDO, the HP Z3801. That is about what you can expect without heroic tuning efforts (on a GPSDO that supports user fiddling, such as the Trimble Thunderbolt). Since many counters gate for one second (or gate faster, then average), the ADEV at tau=1 second is a reasonable estimate of the real-world performance of a good counter driven by a GPSDO. Note that, since the ADEV at 1 second depends largely on the OCXO, the heroic tuning efforts mentioned above won't change the ADEV at tau=1 second hardly at all. That is attributable strictly to the OCXO (unless the GPSDO is very poorly designed). Building a GPSDO using the best OCXO available can improve things. For example, look at the "BVA" plot -- at 1 second, its ADEV is nearly a full decade better than the Z3801, better even than the hydrogen masers. (While I referred to the 3801 above as "decent but not stellar," that is based on its performance at tau from 100-100k seconds. Its OCXO -- and, thus, its ADEV -- is better than most others at 1 second.) Best regards, Charles
BK
Bob kb8tq
Sat, Aug 20, 2016 12:28 PM

Hi

I do not know of any commercial GPSDO's that do not have a holdover
feature. There are a lot of variables. With a large package OCXO and a couple days
operation, the numbers I gave should be good 90 to 99% of the time.(99 seconds out
of 100).

If the unit has such a poor antenna location that (unlike the original post) it never
completes survey, then all bets are off... The basic assumption is "locked to > 4 sats
at all times. That generally translates to 6 or more most of the time.

Bob

Sent from my iPad

On Aug 20, 2016, at 5:43 AM, Bryan _ bpl521@outlook.com wrote:

Bob:
With a GPSDO as a general rule is there much variation of the 0.005 to 0.02 ppb range. Would there me much if only 2 satellites were visible but the unit was no in holdover vs say 6 satellites. I assume there is probably a lot of variables such as quality of GPSDO and OCXO, multi path interference, antenna location, etc, but worse case what would a user expect assuming the unit isn't in holdover.

-=Bryan=-

From: kb8tq@n1k.org
Date: Fri, 19 Aug 2016 20:03:59 -0400
To: time-nuts@febo.com
Subject: Re: [time-nuts] Expected 10 MHz offset from a GPSDO?

Hi

Ok let's toss some numbers into the mix.

The counter time base one day after calibration is in the 0.5 to 1.5 ppb range.

The LPRO ten years after it left the factory is in the 0.5 to 1.5 ppb range.

The GPSDO when running properly should be in the 0.005 to 0.02 ppb range at one second.

One ppb at 10 MHz is 0.01 Hz. It also is a the resolution limit on a 5335 at 1 second.

Bob

On Aug 19, 2016, at 5:29 PM, Mark Sims holrum@hotmail.com wrote:

Try driving your counter frequency reference input with the LPRO or comparing the LPRO with the GPSDO.  The LPRO is (most likely) much more accurate the counters' internal timebase.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi I do not know of any commercial GPSDO's that do not have a holdover feature. There are a lot of variables. With a large package OCXO and a couple days operation, the numbers I gave should be good 90 to 99% of the time.(99 seconds out of 100). If the unit has such a poor antenna location that (unlike the original post) it never completes survey, then all bets are off... The basic assumption is "locked to > 4 sats at all times. That generally translates to 6 or more most of the time. Bob Sent from my iPad > On Aug 20, 2016, at 5:43 AM, Bryan _ <bpl521@outlook.com> wrote: > > Bob: > With a GPSDO as a general rule is there much variation of the 0.005 to 0.02 ppb range. Would there me much if only 2 satellites were visible but the unit was no in holdover vs say 6 satellites. I assume there is probably a lot of variables such as quality of GPSDO and OCXO, multi path interference, antenna location, etc, but worse case what would a user expect assuming the unit isn't in holdover. > > -=Bryan=- > >> From: kb8tq@n1k.org >> Date: Fri, 19 Aug 2016 20:03:59 -0400 >> To: time-nuts@febo.com >> Subject: Re: [time-nuts] Expected 10 MHz offset from a GPSDO? >> >> Hi >> >> Ok let's toss some numbers into the mix. >> >> The counter time base one day after calibration is in the 0.5 to 1.5 ppb range. >> >> The LPRO ten years after it left the factory is in the 0.5 to 1.5 ppb range. >> >> The GPSDO when running properly should be in the 0.005 to 0.02 ppb range at one second. >> >> One ppb at 10 MHz is 0.01 Hz. It also is a the resolution limit on a 5335 at 1 second. >> >> Bob >> >> >> >>> On Aug 19, 2016, at 5:29 PM, Mark Sims <holrum@hotmail.com> wrote: >>> >>> Try driving your counter frequency reference input with the LPRO or comparing the LPRO with the GPSDO. The LPRO is (most likely) much more accurate the counters' internal timebase. >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BK
Bob kb8tq
Sat, Aug 20, 2016 3:14 PM

Hi

At the most basic level, ADEV is a standard deviation. When people read an instrument they usually want a "max error" sort of number. ADEV operates on delta between readings so a simple "three sigma" mental math conversion is not all you need. It is the right measure. It can take a bit of work to convert it to useful units .

Bob

On Aug 20, 2016, at 7:38 AM, Charles Steinmetz csteinmetz@yandex.com wrote:

Bryan wrote:

With a GPSDO as a general rule is there much variation of the 0.005 to 0.02 ppb range. Would there me much if only 2 satellites were visible but the unit was no in holdover vs say 6 satellites. I assume there is probably a lot of variables such as quality of GPSDO and OCXO, multi path interference, antenna location, etc, but worse case what would a user expect assuming the unit isn't in holdover.

There are lots of ADEV plots of GPSDOs on the web, if you search for them.  Tom has the attached plot posted on his web site -- it includes one decent (but not stellar) GPSDO, the HP Z3801.  That is about what you can expect without heroic tuning efforts (on a GPSDO that supports user fiddling, such as the Trimble Thunderbolt).

Since many counters gate for one second (or gate faster, then average), the ADEV at tau=1 second is a reasonable estimate of the real-world performance of a good counter driven by a GPSDO.  Note that, since the ADEV at 1 second depends largely on the OCXO, the heroic tuning efforts mentioned above won't change the ADEV at tau=1 second hardly at all. That is attributable strictly to the OCXO (unless the GPSDO is very poorly designed).  Building a GPSDO using the best OCXO available can improve things.  For example, look at the "BVA" plot -- at 1 second, its ADEV is nearly a full decade better than the Z3801, better even than the hydrogen masers.  (While I referred to the 3801 above as "decent but not stellar," that is based on its performance at tau from 100-100k seconds.  Its OCXO -- and, thus, its ADEV -- is better than most others at 1 second.)

Best regards,

Charles

<ADEV_plots_of_various_oscillators_TVB.gif>


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi At the most basic level, ADEV is a standard deviation. When people read an instrument they usually want a "max error" sort of number. ADEV operates on delta between readings so a simple "three sigma" mental math conversion is not all you need. It *is* the right measure. It can take a bit of work to convert it to useful units . Bob > On Aug 20, 2016, at 7:38 AM, Charles Steinmetz <csteinmetz@yandex.com> wrote: > > Bryan wrote: > >> With a GPSDO as a general rule is there much variation of the 0.005 to 0.02 ppb range. Would there me much if only 2 satellites were visible but the unit was no in holdover vs say 6 satellites. I assume there is probably a lot of variables such as quality of GPSDO and OCXO, multi path interference, antenna location, etc, but worse case what would a user expect assuming the unit isn't in holdover. > > There are lots of ADEV plots of GPSDOs on the web, if you search for them. Tom has the attached plot posted on his web site -- it includes one decent (but not stellar) GPSDO, the HP Z3801. That is about what you can expect without heroic tuning efforts (on a GPSDO that supports user fiddling, such as the Trimble Thunderbolt). > > Since many counters gate for one second (or gate faster, then average), the ADEV at tau=1 second is a reasonable estimate of the real-world performance of a good counter driven by a GPSDO. Note that, since the ADEV at 1 second depends largely on the OCXO, the heroic tuning efforts mentioned above won't change the ADEV at tau=1 second hardly at all. That is attributable strictly to the OCXO (unless the GPSDO is very poorly designed). Building a GPSDO using the best OCXO available can improve things. For example, look at the "BVA" plot -- at 1 second, its ADEV is nearly a full decade better than the Z3801, better even than the hydrogen masers. (While I referred to the 3801 above as "decent but not stellar," that is based on its performance at tau from 100-100k seconds. Its OCXO -- and, thus, its ADEV -- is better than most others at 1 second.) > > Best regards, > > Charles > > > <ADEV_plots_of_various_oscillators_TVB.gif> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.