Re: [time-nuts] uncertainty calculations

J
jimlux
Fri, Apr 14, 2017 3:49 PM

On 4/14/17 8:37 AM, jimlux wrote:

If one is counting an unknown 1pps source with a counter that runs at 10
MHz (e.g. the error in any one measurement is uniformly distributed over
1 ppm) and you collect 100 samples,
is the (1 sigma) measurement uncertainty 0.1ppm * sqrt(100)/sqrt(12)

(standard deviation of a uniform distribution is 1/sqrt(12) )

(assuming for the moment that both sources have no underlying
variability - we're talking about the measurement uncertainty)

Oops..
0.1 ppm * 1/sqrt(N) * 1/sqrt(12)

That is, the standard  deviation goes down as sqrt(N)

On 4/14/17 8:37 AM, jimlux wrote: > If one is counting an unknown 1pps source with a counter that runs at 10 > MHz (e.g. the error in any one measurement is uniformly distributed over > 1 ppm) and you collect 100 samples, > is the (1 sigma) measurement uncertainty 0.1ppm * sqrt(100)/sqrt(12) > > (standard deviation of a uniform distribution is 1/sqrt(12) ) > > (assuming for the moment that both sources have no underlying > variability - we're talking about the *measurement uncertainty*) > Oops.. 0.1 ppm * 1/sqrt(N) * 1/sqrt(12) That is, the standard deviation goes down as sqrt(N)
EA
Ed Armstrong
Sat, Apr 15, 2017 8:13 AM

My NTGS50AA is suddenly lacking the ability to communicate. It's been
working fine for months, but I was trying a few things with my Raspberry
Pi NTP server.

Had GPSD accessing the NTGS50AA through a ch340 USB to serial adapter.
Found out Lady Heather displays much less info when connecting through
GPSD, so I went back to LH connecting directly to NTGS50AA through ch340
adapter on windows 7. Curiously, it was surveying, not in position lock.
So, I did a hard reset. Never heard from it since :(

If it's unplugged, all the lights behave as they should when you power
it up again. "Locked" light eventually comes on. PPS still works. Just
no serial. Even tried Trimble Studio with auto detect settings, no
connect here either. I've tried a second ch340 USB to serial adapter,
same deal. Any suggestions?

Thanks
Ed

My NTGS50AA is suddenly lacking the ability to communicate. It's been working fine for months, but I was trying a few things with my Raspberry Pi NTP server. Had GPSD accessing the NTGS50AA through a ch340 USB to serial adapter. Found out Lady Heather displays much less info when connecting through GPSD, so I went back to LH connecting directly to NTGS50AA through ch340 adapter on windows 7. Curiously, it was surveying, not in position lock. So, I did a hard reset. Never heard from it since :( If it's unplugged, all the lights behave as they should when you power it up again. "Locked" light eventually comes on. PPS still works. Just no serial. Even tried Trimble Studio with auto detect settings, no connect here either. I've tried a second ch340 USB to serial adapter, same deal. Any suggestions? Thanks Ed
B_
Bryan _
Sat, Apr 15, 2017 11:11 AM

Have you tried different serial settings I wonder if the hard reset mucked things up. have a Nortel as well, never had a issue with a reset but try changing a few thing sup. It may need the unusual 19200,7,Odd and 1 stop bit setting.

If you try a terminal program do you at least see some garbage characters?

-=Bryan=-


From: time-nuts time-nuts-bounces@febo.com on behalf of Ed Armstrong eds_equipment@verizon.net
Sent: April 15, 2017 1:13 AM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] NTGS50AA No Serial

My NTGS50AA is suddenly lacking the ability to communicate. It's been
working fine for months, but I was trying a few things with my Raspberry
Pi NTP server.

Had GPSD accessing the NTGS50AA through a ch340 USB to serial adapter.
Found out Lady Heather displays much less info when connecting through
GPSD, so I went back to LH connecting directly to NTGS50AA through ch340
adapter on windows 7. Curiously, it was surveying, not in position lock.
So, I did a hard reset. Never heard from it since :(

If it's unplugged, all the lights behave as they should when you power
it up again. "Locked" light eventually comes on. PPS still works. Just
no serial. Even tried Trimble Studio with auto detect settings, no
connect here either. I've tried a second ch340 USB to serial adapter,
same deal. Any suggestions?

Thanks
Ed


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
time-nuts Info Page - American Febo Enterpriseshttps://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
www.febo.com
time-nuts is a low volume, high SNR list for the discussion of precise time and frequency measurement and related topics. To see the collection of prior postings to ...

and follow the instructions there.

Have you tried different serial settings I wonder if the hard reset mucked things up. have a Nortel as well, never had a issue with a reset but try changing a few thing sup. It may need the unusual 19200,7,Odd and 1 stop bit setting. If you try a terminal program do you at least see some garbage characters? -=Bryan=- ________________________________ From: time-nuts <time-nuts-bounces@febo.com> on behalf of Ed Armstrong <eds_equipment@verizon.net> Sent: April 15, 2017 1:13 AM To: Discussion of precise time and frequency measurement Subject: [time-nuts] NTGS50AA No Serial My NTGS50AA is suddenly lacking the ability to communicate. It's been working fine for months, but I was trying a few things with my Raspberry Pi NTP server. Had GPSD accessing the NTGS50AA through a ch340 USB to serial adapter. Found out Lady Heather displays much less info when connecting through GPSD, so I went back to LH connecting directly to NTGS50AA through ch340 adapter on windows 7. Curiously, it was surveying, not in position lock. So, I did a hard reset. Never heard from it since :( If it's unplugged, all the lights behave as they should when you power it up again. "Locked" light eventually comes on. PPS still works. Just no serial. Even tried Trimble Studio with auto detect settings, no connect here either. I've tried a second ch340 USB to serial adapter, same deal. Any suggestions? Thanks Ed _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts time-nuts Info Page - American Febo Enterprises<https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts> www.febo.com time-nuts is a low volume, high SNR list for the discussion of precise time and frequency measurement and related topics. To see the collection of prior postings to ... and follow the instructions there.
E
EB4APL
Sat, Apr 15, 2017 3:32 PM

I had a stupid problem with mine when doing some tests and I had the
same symptoms and it took me several days of tests to realize the problem.

The original flat cable that goes to the auxiliary board with the panel
lights and the RS-232 connector has the pins reversed so if you connect
the pin one to pin one you still get some LEDs on but no comms. Connect
it as per the attached photo.  If the list deletes the attachments,
please ask me to send it directly.

El 15/04/2017 a las 13:11, Bryan _ escribió:

Have you tried different serial settings I wonder if the hard reset mucked things up. have a Nortel as well, never had a issue with a reset but try changing a few thing sup. It may need the unusual 19200,7,Odd and 1 stop bit setting.

If you try a terminal program do you at least see some garbage characters?

-=Bryan=-


From: time-nuts time-nuts-bounces@febo.com on behalf of Ed Armstrong eds_equipment@verizon.net
Sent: April 15, 2017 1:13 AM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] NTGS50AA No Serial

My NTGS50AA is suddenly lacking the ability to communicate. It's been
working fine for months, but I was trying a few things with my Raspberry
Pi NTP server.

Had GPSD accessing the NTGS50AA through a ch340 USB to serial adapter.
Found out Lady Heather displays much less info when connecting through
GPSD, so I went back to LH connecting directly to NTGS50AA through ch340
adapter on windows 7. Curiously, it was surveying, not in position lock.
So, I did a hard reset. Never heard from it since :(

If it's unplugged, all the lights behave as they should when you power
it up again. "Locked" light eventually comes on. PPS still works. Just
no serial. Even tried Trimble Studio with auto detect settings, no
connect here either. I've tried a second ch340 USB to serial adapter,
same deal. Any suggestions?

Thanks
Ed


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
time-nuts Info Page - American Febo Enterpriseshttps://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
www.febo.com
time-nuts is a low volume, high SNR list for the discussion of precise time and frequency measurement and related topics. To see the collection of prior postings to ...

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

I had a stupid problem with mine when doing some tests and I had the same symptoms and it took me several days of tests to realize the problem. The original flat cable that goes to the auxiliary board with the panel lights and the RS-232 connector has the pins reversed so if you connect the pin one to pin one you still get some LEDs on but no comms. Connect it as per the attached photo. If the list deletes the attachments, please ask me to send it directly. El 15/04/2017 a las 13:11, Bryan _ escribió: > Have you tried different serial settings I wonder if the hard reset mucked things up. have a Nortel as well, never had a issue with a reset but try changing a few thing sup. It may need the unusual 19200,7,Odd and 1 stop bit setting. > > > If you try a terminal program do you at least see some garbage characters? > > -=Bryan=- > > > ________________________________ > From: time-nuts <time-nuts-bounces@febo.com> on behalf of Ed Armstrong <eds_equipment@verizon.net> > Sent: April 15, 2017 1:13 AM > To: Discussion of precise time and frequency measurement > Subject: [time-nuts] NTGS50AA No Serial > > My NTGS50AA is suddenly lacking the ability to communicate. It's been > working fine for months, but I was trying a few things with my Raspberry > Pi NTP server. > > Had GPSD accessing the NTGS50AA through a ch340 USB to serial adapter. > Found out Lady Heather displays much less info when connecting through > GPSD, so I went back to LH connecting directly to NTGS50AA through ch340 > adapter on windows 7. Curiously, it was surveying, not in position lock. > So, I did a hard reset. Never heard from it since :( > > If it's unplugged, all the lights behave as they should when you power > it up again. "Locked" light eventually comes on. PPS still works. Just > no serial. Even tried Trimble Studio with auto detect settings, no > connect here either. I've tried a second ch340 USB to serial adapter, > same deal. Any suggestions? > > Thanks > Ed > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > time-nuts Info Page - American Febo Enterprises<https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts> > www.febo.com > time-nuts is a low volume, high SNR list for the discussion of precise time and frequency measurement and related topics. To see the collection of prior postings to ... > > > > and follow the instructions there. > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
E
EB4APL
Sat, Apr 15, 2017 5:05 PM

I had a stupid problem with mine when doing some tests and I had the
same symptoms and it took me several days of tests to realize the problem.

The original flat cable that goes to the auxiliary board with the panel
lights and the RS-232 connector has the pins reversed so if you connect
the pin one to pin one you still get some LEDs on but no comms. Connect
it as per the attached photo.  If the list deletes the attachments,
please ask me to send it directly.

El 15/04/2017 a las 13:11, Bryan _ escribió:

Have you tried different serial settings I wonder if the hard reset mucked things up. have a Nortel as well, never had a issue with a reset but try changing a few thing sup. It may need the unusual 19200,7,Odd and 1 stop bit setting.

If you try a terminal program do you at least see some garbage characters?

-=Bryan=-


From: time-nuts time-nuts-bounces@febo.com on behalf of Ed Armstrong eds_equipment@verizon.net
Sent: April 15, 2017 1:13 AM
To: Discussion of precise time and frequency measurement
Subject: [time-nuts] NTGS50AA No Serial

My NTGS50AA is suddenly lacking the ability to communicate. It's been
working fine for months, but I was trying a few things with my Raspberry
Pi NTP server.

Had GPSD accessing the NTGS50AA through a ch340 USB to serial adapter.
Found out Lady Heather displays much less info when connecting through
GPSD, so I went back to LH connecting directly to NTGS50AA through ch340
adapter on windows 7. Curiously, it was surveying, not in position lock.
So, I did a hard reset. Never heard from it since :(

If it's unplugged, all the lights behave as they should when you power
it up again. "Locked" light eventually comes on. PPS still works. Just
no serial. Even tried Trimble Studio with auto detect settings, no
connect here either. I've tried a second ch340 USB to serial adapter,
same deal. Any suggestions?

Thanks
Ed


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
time-nuts Info Page - American Febo Enterpriseshttps://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
www.febo.com
time-nuts is a low volume, high SNR list for the discussion of precise time and frequency measurement and related topics. To see the collection of prior postings to ...

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

I had a stupid problem with mine when doing some tests and I had the same symptoms and it took me several days of tests to realize the problem. The original flat cable that goes to the auxiliary board with the panel lights and the RS-232 connector has the pins reversed so if you connect the pin one to pin one you still get some LEDs on but no comms. Connect it as per the attached photo. If the list deletes the attachments, please ask me to send it directly. El 15/04/2017 a las 13:11, Bryan _ escribió: > Have you tried different serial settings I wonder if the hard reset mucked things up. have a Nortel as well, never had a issue with a reset but try changing a few thing sup. It may need the unusual 19200,7,Odd and 1 stop bit setting. > > > If you try a terminal program do you at least see some garbage characters? > > -=Bryan=- > > > ________________________________ > From: time-nuts <time-nuts-bounces@febo.com> on behalf of Ed Armstrong <eds_equipment@verizon.net> > Sent: April 15, 2017 1:13 AM > To: Discussion of precise time and frequency measurement > Subject: [time-nuts] NTGS50AA No Serial > > My NTGS50AA is suddenly lacking the ability to communicate. It's been > working fine for months, but I was trying a few things with my Raspberry > Pi NTP server. > > Had GPSD accessing the NTGS50AA through a ch340 USB to serial adapter. > Found out Lady Heather displays much less info when connecting through > GPSD, so I went back to LH connecting directly to NTGS50AA through ch340 > adapter on windows 7. Curiously, it was surveying, not in position lock. > So, I did a hard reset. Never heard from it since :( > > If it's unplugged, all the lights behave as they should when you power > it up again. "Locked" light eventually comes on. PPS still works. Just > no serial. Even tried Trimble Studio with auto detect settings, no > connect here either. I've tried a second ch340 USB to serial adapter, > same deal. Any suggestions? > > Thanks > Ed > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > time-nuts Info Page - American Febo Enterprises<https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts> > www.febo.com > time-nuts is a low volume, high SNR list for the discussion of precise time and frequency measurement and related topics. To see the collection of prior postings to ... > > > > and follow the instructions there. > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
BB
Bill Byrom
Sun, Apr 16, 2017 1:35 AM

I believe that the problem is that the error in any one measurement is
not uniformly distributed in exactly that way. If you are trying to
count how many 10 MHz (100 ns interval) intervals occur between the 1
PPS edges in a period counter, you have to deal with the following?

  • You might have any possible phase relationship between the two
    signals. If they are exactly related by a 10^7 ratio, it's possible
    for the 1 PPS edges to exactly coincide with the 10 MHz edges.
    Depending on the type of gating circuit, you will have jitter and
    possibly metastability resolving whether which edge occured first. The
    same thing happens on the end of the measured interval, but (depending
    on how it's set up) the propagation delays and metastability and
    jitter might be different. So you could get millions of sequential
    counts which were 1 count low, followed by millions of counts which
    were one count high, with no counts exactly at 10^7.

  • To stay away from such problems, most precision counters add a small
    amount of controlled jitter (phase modulation) to the clock. When
    averaged over many measurements the effects of the two edges (gate and
    clock) lining up exactly are greatly reduced, since you are sliding
    one back and forth across the other with the modulation and the chance
    of metastability is small (assuming the signal being measured doesn't
    happen to match the phase modulation frequency).

  • The metastability problem depends on how the edges are compared. Some
    traditional flip-flops and latches can be thought of as analog gain
    elements connected so that they tend to sit in state A or state B,
    which involve analog voltages and currents. If you graph the energy in
    the system, the energy is low in state A, rises to a peak halfway
    between A and B, and falls to a low value at state B. If the
    recognition of the timing edge occurs early enough the system remains
    in state A. If the timing is later the system is pushed toward the
    peak, but doesn't get over it and returns to state A. But if the
    timing is at the perfect location the system is balanced at the
    potential energy peak, and only random noise can push the system into
    a final state A or B over a significant length of time.

Sorry if this is considered obvious or trivial.

--

Bill Byrom N5BB

----- Original message -----

From: jimlux jimlux@earthlink.net

To: Discussion of precise time and frequency measurement time-nuts@febo.com
Subject: Re: [time-nuts] uncertainty calculations

Date: Fri, 14 Apr 2017 08:49:07 -0700

On 4/14/17 8:37 AM, jimlux wrote:

If one is counting an unknown 1pps source with a counter that
runs at 10
MHz (e.g. the error in any one measurement is uniformly
distributed over
1 ppm) and you collect 100 samples,

is the (1 sigma) measurement uncertainty 0.1ppm * sqrt(100)/sqrt(12)

(standard deviation of a uniform distribution is 1/sqrt(12) )

(assuming for the moment that both sources have no underlying

variability - we're talking about the measurement uncertainty)

Oops..

0.1 ppm * 1/sqrt(N) * 1/sqrt(12)

That is, the standard  deviation goes down as sqrt(N)


time-nuts mailing list -- time-nuts@febo.com

To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

I believe that the problem is that the error in any one measurement is not uniformly distributed in exactly that way. If you are trying to count how many 10 MHz (100 ns interval) intervals occur between the 1 PPS edges in a period counter, you have to deal with the following? * You might have any possible phase relationship between the two signals. If they are exactly related by a 10^7 ratio, it's possible for the 1 PPS edges to exactly coincide with the 10 MHz edges. Depending on the type of gating circuit, you will have jitter and possibly metastability resolving whether which edge occured first. The same thing happens on the end of the measured interval, but (depending on how it's set up) the propagation delays and metastability and jitter might be different. So you could get millions of sequential counts which were 1 count low, followed by millions of counts which were one count high, with no counts exactly at 10^7. * To stay away from such problems, most precision counters add a small amount of controlled jitter (phase modulation) to the clock. When averaged over many measurements the effects of the two edges (gate and clock) lining up exactly are greatly reduced, since you are sliding one back and forth across the other with the modulation and the chance of metastability is small (assuming the signal being measured doesn't happen to match the phase modulation frequency). * The metastability problem depends on how the edges are compared. Some traditional flip-flops and latches can be thought of as analog gain elements connected so that they tend to sit in state A or state B, which involve analog voltages and currents. If you graph the energy in the system, the energy is low in state A, rises to a peak halfway between A and B, and falls to a low value at state B. If the recognition of the timing edge occurs early enough the system remains in state A. If the timing is later the system is pushed toward the peak, but doesn't get over it and returns to state A. But if the timing is at the perfect location the system is balanced at the potential energy peak, and only random noise can push the system into a final state A or B over a significant length of time. Sorry if this is considered obvious or trivial. -- Bill Byrom N5BB ----- Original message ----- From: jimlux <jimlux@earthlink.net> To: Discussion of precise time and frequency measurement <time-nuts@febo.com> Subject: Re: [time-nuts] uncertainty calculations Date: Fri, 14 Apr 2017 08:49:07 -0700 On 4/14/17 8:37 AM, jimlux wrote: > If one is counting an unknown 1pps source with a counter that > runs at 10 > MHz (e.g. the error in any one measurement is uniformly > distributed over > 1 ppm) and you collect 100 samples, > is the (1 sigma) measurement uncertainty 0.1ppm * sqrt(100)/sqrt(12) > > (standard deviation of a uniform distribution is 1/sqrt(12) ) > > (assuming for the moment that both sources have no underlying > variability - we're talking about the *measurement uncertainty*) > Oops.. 0.1 ppm * 1/sqrt(N) * 1/sqrt(12) That is, the standard deviation goes down as sqrt(N) _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.