volt-nuts@lists.febo.com

Discussion of precise voltage measurement

View all threads

Vibrating reed electrometers

G
george
Sat, Mar 3, 2018 5:34 PM

Hi all

To the best of my memory Keithley never made vibrating reed electrometers, the only one that I am aware of is the Varian Cary 401 which did use Sapphire insulators.  I was the European product line specialist for Varian Cary in the late 1960/1970 era and was involved with the 401.

George G6HIG


From: volt-nuts volt-nuts-bounces@febo.com on behalf of volt-nuts-request@febo.com volt-nuts-request@febo.com
Sent: 03 March 2018 17:00
To: volt-nuts@febo.com
Subject: volt-nuts Digest, Vol 103, Issue 3

Send volt-nuts mailing list submissions to
volt-nuts@febo.com

To subscribe or unsubscribe via the World Wide Web, visit
https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.febo.com%2Fcgi-bin%2Fmailman%2Flistinfo%2Fvolt-nuts&data=02%7C01%7C%7Cb5818846d36e4ecd914908d581283c52%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C636556932161076932&sdata=cp01qd%2FkRw4yYSLYOqJ8RwUF8oOxtdqLVOmDal7gBhs%3D&reserved=0
or, via email, send a message with subject or body 'help' to
volt-nuts-request@febo.com

You can reach the person managing the list at
volt-nuts-owner@febo.com

When replying, please edit your Subject line so it is more specific
than "Re: Contents of volt-nuts digest..."

Today's Topics:

  1. Re: Precision high resistance measurements / calibration of
    HP 4339B high-resistance meter. (ed breya)
  2. Re: Precision high resistance measurements /      calibration of
    HP 4339B high-resistance meter. (Mitch Van Ochten)

Message: 1
Date: Fri, 2 Mar 2018 09:22:32 -0800
From: ed breya eb@telight.com
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Precision high resistance measurements /
calibration of HP 4339B high-resistance meter.
Message-ID: 5f6e418c-649b-70e8-8502-facf1176f390@telight.com
Content-Type: text/plain; charset=utf-8; format=flowed

Oops - I think I didn't send this message properly yesterday - here goes
again. Ed

Yes, David, unless you go to very extreme measures, you won't see real R
values that have any practical meaning beyond E12 ohms or so. Most
practical insulation Rs may be around E12-E14 tops, unless you go to
sapphire. Up in that region, the R may be all within a material, or
include surface components like a film of dirt or moisture, or a
fingerprint.

E11 resistors can be made to fairly high precision, and maybe E12
nowadays. In the old days, higher values were made by stacking E11s -
like ten in series to get E12 with decent precision. The glass packaging
also limits how high it can go, due to leakage within and on the
surface. I once used a glass reed relay capsule as an ultra-high
resistance in a circuit. There was no precision or stability at all, but
it made a nice high resistor (probably E14-ish dry) even though there
was no element in there, and the circuit didn't care, as long as it was
very high, but not infinite.

The specs on this HP unit are likely just the most extreme capability
taking maximum voltage over minimum current resolution, but any
measurements would tend to be very noisy and unstable anyway. Also,
testing at the extreme 1 kV makes the numbers seem more impressive, but
the voltage coefficient of resistance will pretty much be unpredictable.

If this is a digital meter, then the other spec trick that tends to
obscure the real performance limit is that the ultimate resolution and
noise is that last digit - or even last two or three - that may may be
pretty jumpy, unless very long averaging time is used.

There may be newer, fancier electrometers nowadays, but Keithley used to
be the standard for these in the old days, before several digits of DVM
resolution complicated the specs. They had a vibrating capacitor
electrometer with all-sapphire input structure back in the 1970s/80s I
think, that was the epitome of electrometers. I forget the model number,
but vaguely recall that it could reach the aA region full scale - not
that last digit of resolution thing. It's long obsolete, and I don't
think they ever made anything actually better - only added DVM digits to
less capable, conventional semiconductor amplifier techniques. If you
can find info on it, it's an interesting read. I found a pdf of the
manual years ago, but have no idea where it is now, or what info may
still be around.

Ed


Message: 2
Date: Fri, 2 Mar 2018 20:41:10 -0500
From: "Mitch Van Ochten" Mitch@vincentelectronics.com
To: "'Discussion of precise voltage measurement'" volt-nuts@febo.com
Subject: Re: [volt-nuts] Precision high resistance measurements /
calibration of HP 4339B high-resistance meter.
Message-ID: 014e01d3b290$b54948f0$1fdbdad0$@vincentelectronics.com
Content-Type: text/plain;      charset="UTF-8"

Here are specs from an older General Radio Bridge for observation.

GR 1644-A specifications:

Resistance Range: 1 kΩ to 1000 TΩ (10^3 to 10^15 Ω) in ten ranges.
Accuracy: 10^3 Ω to 10^10 Ω, ±1 %. After self-calibration: 10^10 to
10^12 Ω, ±1%*; 10^13 Ω, ±2%; 10^14 Ω, ±10%; 10^15 Ω, ± one scale
division.
ΔR% Dial: ±5% range; accurate to ±0.2% or, for small changes,
to ±0.1%.
Test Voltage: Voltage accuracy is ±3% ±0.5 V.
Fixed Voltages** 10 20 50 100 200 500 1000 v

Best regards,

mitch

-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of ed breya
Sent: Friday, March 02, 2018 12:23 PM
To: volt-nuts@febo.com
Subject: Re: [volt-nuts] Precision high resistance measurements / calibration of HP 4339B high-resistance meter.

Oops - I think I didn't send this message properly yesterday - here goes again. Ed

Yes, David, unless you go to very extreme measures, you won't see real R values that have any practical meaning beyond E12 ohms or so. Most practical insulation Rs may be around E12-E14 tops, unless you go to sapphire. Up in that region, the R may be all within a material, or include surface components like a film of dirt or moisture, or a fingerprint.

E11 resistors can be made to fairly high precision, and maybe E12 nowadays. In the old days, higher values were made by stacking E11s - like ten in series to get E12 with decent precision. The glass packaging also limits how high it can go, due to leakage within and on the surface. I once used a glass reed relay capsule as an ultra-high resistance in a circuit. There was no precision or stability at all, but it made a nice high resistor (probably E14-ish dry) even though there was no element in there, and the circuit didn't care, as long as it was very high, but not infinite.

The specs on this HP unit are likely just the most extreme capability taking maximum voltage over minimum current resolution, but any measurements would tend to be very noisy and unstable anyway. Also, testing at the extreme 1 kV makes the numbers seem more impressive, but the voltage coefficient of resistance will pretty much be unpredictable.

If this is a digital meter, then the other spec trick that tends to obscure the real performance limit is that the ultimate resolution and noise is that last digit - or even last two or three - that may may be pretty jumpy, unless very long averaging time is used.

There may be newer, fancier electrometers nowadays, but Keithley used to be the standard for these in the old days, before several digits of DVM resolution complicated the specs. They had a vibrating capacitor electrometer with all-sapphire input structure back in the 1970s/80s I think, that was the epitome of electrometers. I forget the model number, but vaguely recall that it could reach the aA region full scale - not that last digit of resolution thing. It's long obsolete, and I don't think they ever made anything actually better - only added DVM digits to less capable, conventional semiconductor amplifier techniques. If you can find info on it, it's an interesting read. I found a pdf of the manual years ago, but have no idea where it is now, or what info may still be around.

Ed


volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.febo.com%2Fcgi-bin%2Fmailman%2Flistinfo%2Fvolt-nuts&data=02%7C01%7C%7Cb5818846d36e4ecd914908d581283c52%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C636556932161076932&sdata=cp01qd%2FkRw4yYSLYOqJ8RwUF8oOxtdqLVOmDal7gBhs%3D&reserved=0
and follow the instructions there.


Subject: Digest Footer


volt-nuts mailing list
volt-nuts@febo.com
https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.febo.com%2Fcgi-bin%2Fmailman%2Flistinfo%2Fvolt-nuts&data=02%7C01%7C%7Cb5818846d36e4ecd914908d581283c52%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C636556932161076932&sdata=cp01qd%2FkRw4yYSLYOqJ8RwUF8oOxtdqLVOmDal7gBhs%3D&reserved=0


End of volt-nuts Digest, Vol 103, Issue 3


Hi all To the best of my memory Keithley never made vibrating reed electrometers, the only one that I am aware of is the Varian Cary 401 which did use Sapphire insulators. I was the European product line specialist for Varian Cary in the late 1960/1970 era and was involved with the 401. George G6HIG ________________________________________ From: volt-nuts <volt-nuts-bounces@febo.com> on behalf of volt-nuts-request@febo.com <volt-nuts-request@febo.com> Sent: 03 March 2018 17:00 To: volt-nuts@febo.com Subject: volt-nuts Digest, Vol 103, Issue 3 Send volt-nuts mailing list submissions to volt-nuts@febo.com To subscribe or unsubscribe via the World Wide Web, visit https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.febo.com%2Fcgi-bin%2Fmailman%2Flistinfo%2Fvolt-nuts&data=02%7C01%7C%7Cb5818846d36e4ecd914908d581283c52%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C636556932161076932&sdata=cp01qd%2FkRw4yYSLYOqJ8RwUF8oOxtdqLVOmDal7gBhs%3D&reserved=0 or, via email, send a message with subject or body 'help' to volt-nuts-request@febo.com You can reach the person managing the list at volt-nuts-owner@febo.com When replying, please edit your Subject line so it is more specific than "Re: Contents of volt-nuts digest..." Today's Topics: 1. Re: Precision high resistance measurements / calibration of HP 4339B high-resistance meter. (ed breya) 2. Re: Precision high resistance measurements / calibration of HP 4339B high-resistance meter. (Mitch Van Ochten) ---------------------------------------------------------------------- Message: 1 Date: Fri, 2 Mar 2018 09:22:32 -0800 From: ed breya <eb@telight.com> To: volt-nuts@febo.com Subject: Re: [volt-nuts] Precision high resistance measurements / calibration of HP 4339B high-resistance meter. Message-ID: <5f6e418c-649b-70e8-8502-facf1176f390@telight.com> Content-Type: text/plain; charset=utf-8; format=flowed Oops - I think I didn't send this message properly yesterday - here goes again. Ed Yes, David, unless you go to very extreme measures, you won't see real R values that have any practical meaning beyond E12 ohms or so. Most practical insulation Rs may be around E12-E14 tops, unless you go to sapphire. Up in that region, the R may be all within a material, or include surface components like a film of dirt or moisture, or a fingerprint. E11 resistors can be made to fairly high precision, and maybe E12 nowadays. In the old days, higher values were made by stacking E11s - like ten in series to get E12 with decent precision. The glass packaging also limits how high it can go, due to leakage within and on the surface. I once used a glass reed relay capsule as an ultra-high resistance in a circuit. There was no precision or stability at all, but it made a nice high resistor (probably E14-ish dry) even though there was no element in there, and the circuit didn't care, as long as it was very high, but not infinite. The specs on this HP unit are likely just the most extreme capability taking maximum voltage over minimum current resolution, but any measurements would tend to be very noisy and unstable anyway. Also, testing at the extreme 1 kV makes the numbers seem more impressive, but the voltage coefficient of resistance will pretty much be unpredictable. If this is a digital meter, then the other spec trick that tends to obscure the real performance limit is that the ultimate resolution and noise is that last digit - or even last two or three - that may may be pretty jumpy, unless very long averaging time is used. There may be newer, fancier electrometers nowadays, but Keithley used to be the standard for these in the old days, before several digits of DVM resolution complicated the specs. They had a vibrating capacitor electrometer with all-sapphire input structure back in the 1970s/80s I think, that was the epitome of electrometers. I forget the model number, but vaguely recall that it could reach the aA region full scale - not that last digit of resolution thing. It's long obsolete, and I don't think they ever made anything actually better - only added DVM digits to less capable, conventional semiconductor amplifier techniques. If you can find info on it, it's an interesting read. I found a pdf of the manual years ago, but have no idea where it is now, or what info may still be around. Ed ------------------------------ Message: 2 Date: Fri, 2 Mar 2018 20:41:10 -0500 From: "Mitch Van Ochten" <Mitch@vincentelectronics.com> To: "'Discussion of precise voltage measurement'" <volt-nuts@febo.com> Subject: Re: [volt-nuts] Precision high resistance measurements / calibration of HP 4339B high-resistance meter. Message-ID: <014e01d3b290$b54948f0$1fdbdad0$@vincentelectronics.com> Content-Type: text/plain; charset="UTF-8" Here are specs from an older General Radio Bridge for observation. GR 1644-A specifications: Resistance Range: 1 kΩ to 1000 TΩ (10^3 to 10^15 Ω) in ten ranges. Accuracy: 10^3 Ω to 10^10 Ω, ±1 %. After self-calibration: 10^10 to 10^12 Ω, ±1%*; 10^13 Ω, ±2%; 10^14 Ω, ±10%; 10^15 Ω, ± one scale division. ΔR% Dial: ±5% range; accurate to ±0.2% or, for small changes, to ±0.1%. Test Voltage: Voltage accuracy is ±3% ±0.5 V. Fixed Voltages** 10 20 50 100 200 500 1000 v Best regards, mitch -----Original Message----- From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of ed breya Sent: Friday, March 02, 2018 12:23 PM To: volt-nuts@febo.com Subject: Re: [volt-nuts] Precision high resistance measurements / calibration of HP 4339B high-resistance meter. Oops - I think I didn't send this message properly yesterday - here goes again. Ed Yes, David, unless you go to very extreme measures, you won't see real R values that have any practical meaning beyond E12 ohms or so. Most practical insulation Rs may be around E12-E14 tops, unless you go to sapphire. Up in that region, the R may be all within a material, or include surface components like a film of dirt or moisture, or a fingerprint. E11 resistors can be made to fairly high precision, and maybe E12 nowadays. In the old days, higher values were made by stacking E11s - like ten in series to get E12 with decent precision. The glass packaging also limits how high it can go, due to leakage within and on the surface. I once used a glass reed relay capsule as an ultra-high resistance in a circuit. There was no precision or stability at all, but it made a nice high resistor (probably E14-ish dry) even though there was no element in there, and the circuit didn't care, as long as it was very high, but not infinite. The specs on this HP unit are likely just the most extreme capability taking maximum voltage over minimum current resolution, but any measurements would tend to be very noisy and unstable anyway. Also, testing at the extreme 1 kV makes the numbers seem more impressive, but the voltage coefficient of resistance will pretty much be unpredictable. If this is a digital meter, then the other spec trick that tends to obscure the real performance limit is that the ultimate resolution and noise is that last digit - or even last two or three - that may may be pretty jumpy, unless very long averaging time is used. There may be newer, fancier electrometers nowadays, but Keithley used to be the standard for these in the old days, before several digits of DVM resolution complicated the specs. They had a vibrating capacitor electrometer with all-sapphire input structure back in the 1970s/80s I think, that was the epitome of electrometers. I forget the model number, but vaguely recall that it could reach the aA region full scale - not that last digit of resolution thing. It's long obsolete, and I don't think they ever made anything actually better - only added DVM digits to less capable, conventional semiconductor amplifier techniques. If you can find info on it, it's an interesting read. I found a pdf of the manual years ago, but have no idea where it is now, or what info may still be around. Ed _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.febo.com%2Fcgi-bin%2Fmailman%2Flistinfo%2Fvolt-nuts&data=02%7C01%7C%7Cb5818846d36e4ecd914908d581283c52%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C636556932161076932&sdata=cp01qd%2FkRw4yYSLYOqJ8RwUF8oOxtdqLVOmDal7gBhs%3D&reserved=0 and follow the instructions there. ------------------------------ Subject: Digest Footer _______________________________________________ volt-nuts mailing list volt-nuts@febo.com https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.febo.com%2Fcgi-bin%2Fmailman%2Flistinfo%2Fvolt-nuts&data=02%7C01%7C%7Cb5818846d36e4ecd914908d581283c52%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C636556932161076932&sdata=cp01qd%2FkRw4yYSLYOqJ8RwUF8oOxtdqLVOmDal7gBhs%3D&reserved=0 ------------------------------ End of volt-nuts Digest, Vol 103, Issue 3 *****************************************
BG
Bruce Griffiths
Sat, Mar 3, 2018 9:14 PM

The Keithley model 640 was a vibrating capacitor electrometer.

It was available in the 1970's.

Bruce

 On 04 March 2018 at 06:34 george <g_einst@hotmail.com> wrote:

 Hi all

 To the best of my memory Keithley never made vibrating reed electrometers, the only one that I am aware of is the Varian Cary 401 which did use Sapphire insulators. I was the European product line specialist for Varian Cary in the late 1960/1970 era and was involved with the 401.

 George G6HIG

 ________________________________________
 From: volt-nuts <volt-nuts-bounces@febo.com> on behalf of volt-nuts-request@febo.com <volt-nuts-request@febo.com>
 Sent: 03 March 2018 17:00
 To: volt-nuts@febo.com
 Subject: volt-nuts Digest, Vol 103, Issue 3

 Send volt-nuts mailing list submissions to
 volt-nuts@febo.com

 To subscribe or unsubscribe via the World Wide Web, visit
 https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.febo.com%2Fcgi-bin%2Fmailman%2Flistinfo%2Fvolt-nuts&data=02%7C01%7C%7Cb5818846d36e4ecd914908d581283c52%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C636556932161076932&sdata=cp01qd%2FkRw4yYSLYOqJ8RwUF8oOxtdqLVOmDal7gBhs%3D&reserved=0 https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.febo.com%2Fcgi-bin%2Fmailman%2Flistinfo%2Fvolt-nuts&data=02%7C01%7C%7Cb5818846d36e4ecd914908d581283c52%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C636556932161076932&sdata=cp01qd%2FkRw4yYSLYOqJ8RwUF8oOxtdqLVOmDal7gBhs%3D&reserved=0
 or, via email, send a message with subject or body 'help' to
 volt-nuts-request@febo.com

 You can reach the person managing the list at
 volt-nuts-owner@febo.com

 When replying, please edit your Subject line so it is more specific
 than "Re: Contents of volt-nuts digest..."

 Today's Topics:

    1. Re: Precision high resistance measurements / calibration of
       HP 4339B high-resistance meter. (ed breya)
    2. Re: Precision high resistance measurements / calibration of
       HP 4339B high-resistance meter. (Mitch Van Ochten)

 ----------------------------------------------------------------------

 Message: 1
 Date: Fri, 2 Mar 2018 09:22:32 -0800
 From: ed breya <eb@telight.com>
 To: volt-nuts@febo.com
 Subject: Re: [volt-nuts] Precision high resistance measurements /
 calibration of HP 4339B high-resistance meter.
 Message-ID: <5f6e418c-649b-70e8-8502-facf1176f390@telight.com>
 Content-Type: text/plain; charset=utf-8; format=flowed

 Oops - I think I didn't send this message properly yesterday - here goes
 again. Ed

 Yes, David, unless you go to very extreme measures, you won't see real R
 values that have any practical meaning beyond E12 ohms or so. Most
 practical insulation Rs may be around E12-E14 tops, unless you go to
 sapphire. Up in that region, the R may be all within a material, or
 include surface components like a film of dirt or moisture, or a
 fingerprint.

 E11 resistors can be made to fairly high precision, and maybe E12
 nowadays. In the old days, higher values were made by stacking E11s -
 like ten in series to get E12 with decent precision. The glass packaging
 also limits how high it can go, due to leakage within and on the
 surface. I once used a glass reed relay capsule as an ultra-high
 resistance in a circuit. There was no precision or stability at all, but
 it made a nice high resistor (probably E14-ish dry) even though there
 was no element in there, and the circuit didn't care, as long as it was
 very high, but not infinite.

 The specs on this HP unit are likely just the most extreme capability
 taking maximum voltage over minimum current resolution, but any
 measurements would tend to be very noisy and unstable anyway. Also,
 testing at the extreme 1 kV makes the numbers seem more impressive, but
 the voltage coefficient of resistance will pretty much be unpredictable.

 If this is a digital meter, then the other spec trick that tends to
 obscure the real performance limit is that the ultimate resolution and
 noise is that last digit - or even last two or three - that may may be
 pretty jumpy, unless very long averaging time is used.

 There may be newer, fancier electrometers nowadays, but Keithley used to
 be the standard for these in the old days, before several digits of DVM
 resolution complicated the specs. They had a vibrating capacitor
 electrometer with all-sapphire input structure back in the 1970s/80s I
 think, that was the epitome of electrometers. I forget the model number,
 but vaguely recall that it could reach the aA region full scale - not
 that last digit of resolution thing. It's long obsolete, and I don't
 think they ever made anything actually better - only added DVM digits to
 less capable, conventional semiconductor amplifier techniques. If you
 can find info on it, it's an interesting read. I found a pdf of the
 manual years ago, but have no idea where it is now, or what info may
 still be around.

 Ed

 ------------------------------

 Message: 2
 Date: Fri, 2 Mar 2018 20:41:10 -0500
 From: "Mitch Van Ochten" <Mitch@vincentelectronics.com>
 To: "'Discussion of precise voltage measurement'" <volt-nuts@febo.com>
 Subject: Re: [volt-nuts] Precision high resistance measurements /
 calibration of HP 4339B high-resistance meter.
 Message-ID: <014e01d3b290$b54948f0$1fdbdad0$@vincentelectronics.com>
 Content-Type: text/plain; charset="UTF-8"

 Here are specs from an older General Radio Bridge for observation.

 GR 1644-A specifications:

 Resistance Range: 1 kΩ to 1000 TΩ (10^3 to 10^15 Ω) in ten ranges.
 Accuracy: 10^3 Ω to 10^10 Ω, ±1 %. After self-calibration: 10^10 to
 10^12 Ω, ±1%*; 10^13 Ω, ±2%; 10^14 Ω, ±10%; 10^15 Ω, ± one scale
 division.
 ΔR% Dial: ±5% range; accurate to ±0.2% or, for small changes,
 to ±0.1%.
 Test Voltage: Voltage accuracy is ±3% ±0.5 V.
 Fixed Voltages** 10 20 50 100 200 500 1000 v

 Best regards,

 mitch

 -----Original Message-----
 From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of ed breya
 Sent: Friday, March 02, 2018 12:23 PM
 To: volt-nuts@febo.com
 Subject: Re: [volt-nuts] Precision high resistance measurements / calibration of HP 4339B high-resistance meter.

 Oops - I think I didn't send this message properly yesterday - here goes again. Ed

 Yes, David, unless you go to very extreme measures, you won't see real R values that have any practical meaning beyond E12 ohms or so. Most practical insulation Rs may be around E12-E14 tops, unless you go to sapphire. Up in that region, the R may be all within a material, or include surface components like a film of dirt or moisture, or a fingerprint.

 E11 resistors can be made to fairly high precision, and maybe E12 nowadays. In the old days, higher values were made by stacking E11s - like ten in series to get E12 with decent precision. The glass packaging also limits how high it can go, due to leakage within and on the surface. I once used a glass reed relay capsule as an ultra-high resistance in a circuit. There was no precision or stability at all, but it made a nice high resistor (probably E14-ish dry) even though there was no element in there, and the circuit didn't care, as long as it was very high, but not infinite.

 The specs on this HP unit are likely just the most extreme capability taking maximum voltage over minimum current resolution, but any measurements would tend to be very noisy and unstable anyway. Also, testing at the extreme 1 kV makes the numbers seem more impressive, but the voltage coefficient of resistance will pretty much be unpredictable.

 If this is a digital meter, then the other spec trick that tends to obscure the real performance limit is that the ultimate resolution and noise is that last digit - or even last two or three - that may may be pretty jumpy, unless very long averaging time is used.

 There may be newer, fancier electrometers nowadays, but Keithley used to be the standard for these in the old days, before several digits of DVM resolution complicated the specs. They had a vibrating capacitor electrometer with all-sapphire input structure back in the 1970s/80s I think, that was the epitome of electrometers. I forget the model number, but vaguely recall that it could reach the aA region full scale - not that last digit of resolution thing. It's long obsolete, and I don't think they ever made anything actually better - only added DVM digits to less capable, conventional semiconductor amplifier techniques. If you can find info on it, it's an interesting read. I found a pdf of the manual years ago, but have no idea where it is now, or what info may still be around.

 Ed

 _______________________________________________
 volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.febo.com%2Fcgi-bin%2Fmailman%2Flistinfo%2Fvolt-nuts&data=02%7C01%7C%7Cb5818846d36e4ecd914908d581283c52%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C636556932161076932&sdata=cp01qd%2FkRw4yYSLYOqJ8RwUF8oOxtdqLVOmDal7gBhs%3D&reserved=0 https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.febo.com%2Fcgi-bin%2Fmailman%2Flistinfo%2Fvolt-nuts&data=02%7C01%7C%7Cb5818846d36e4ecd914908d581283c52%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C636556932161076932&sdata=cp01qd%2FkRw4yYSLYOqJ8RwUF8oOxtdqLVOmDal7gBhs%3D&reserved=0
 and follow the instructions there.

 ------------------------------

 Subject: Digest Footer

 _______________________________________________
 volt-nuts mailing list
 volt-nuts@febo.com
 https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.febo.com%2Fcgi-bin%2Fmailman%2Flistinfo%2Fvolt-nuts&data=02%7C01%7C%7Cb5818846d36e4ecd914908d581283c52%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C636556932161076932&sdata=cp01qd%2FkRw4yYSLYOqJ8RwUF8oOxtdqLVOmDal7gBhs%3D&reserved=0 https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.febo.com%2Fcgi-bin%2Fmailman%2Flistinfo%2Fvolt-nuts&data=02%7C01%7C%7Cb5818846d36e4ecd914908d581283c52%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C636556932161076932&sdata=cp01qd%2FkRw4yYSLYOqJ8RwUF8oOxtdqLVOmDal7gBhs%3D&reserved=0

 ------------------------------

 End of volt-nuts Digest, Vol 103, Issue 3

 *****************************************

 _______________________________________________
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
The Keithley model 640 was a vibrating capacitor electrometer. It was available in the 1970's. Bruce > > On 04 March 2018 at 06:34 george <g_einst@hotmail.com> wrote: > > Hi all > > To the best of my memory Keithley never made vibrating reed electrometers, the only one that I am aware of is the Varian Cary 401 which did use Sapphire insulators. I was the European product line specialist for Varian Cary in the late 1960/1970 era and was involved with the 401. > > George G6HIG > > ________________________________________ > From: volt-nuts <volt-nuts-bounces@febo.com> on behalf of volt-nuts-request@febo.com <volt-nuts-request@febo.com> > Sent: 03 March 2018 17:00 > To: volt-nuts@febo.com > Subject: volt-nuts Digest, Vol 103, Issue 3 > > Send volt-nuts mailing list submissions to > volt-nuts@febo.com > > To subscribe or unsubscribe via the World Wide Web, visit > https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.febo.com%2Fcgi-bin%2Fmailman%2Flistinfo%2Fvolt-nuts&data=02%7C01%7C%7Cb5818846d36e4ecd914908d581283c52%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C636556932161076932&sdata=cp01qd%2FkRw4yYSLYOqJ8RwUF8oOxtdqLVOmDal7gBhs%3D&reserved=0 https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.febo.com%2Fcgi-bin%2Fmailman%2Flistinfo%2Fvolt-nuts&data=02%7C01%7C%7Cb5818846d36e4ecd914908d581283c52%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C636556932161076932&sdata=cp01qd%2FkRw4yYSLYOqJ8RwUF8oOxtdqLVOmDal7gBhs%3D&reserved=0 > or, via email, send a message with subject or body 'help' to > volt-nuts-request@febo.com > > You can reach the person managing the list at > volt-nuts-owner@febo.com > > When replying, please edit your Subject line so it is more specific > than "Re: Contents of volt-nuts digest..." > > Today's Topics: > > 1. Re: Precision high resistance measurements / calibration of > HP 4339B high-resistance meter. (ed breya) > 2. Re: Precision high resistance measurements / calibration of > HP 4339B high-resistance meter. (Mitch Van Ochten) > > ---------------------------------------------------------------------- > > Message: 1 > Date: Fri, 2 Mar 2018 09:22:32 -0800 > From: ed breya <eb@telight.com> > To: volt-nuts@febo.com > Subject: Re: [volt-nuts] Precision high resistance measurements / > calibration of HP 4339B high-resistance meter. > Message-ID: <5f6e418c-649b-70e8-8502-facf1176f390@telight.com> > Content-Type: text/plain; charset=utf-8; format=flowed > > Oops - I think I didn't send this message properly yesterday - here goes > again. Ed > > Yes, David, unless you go to very extreme measures, you won't see real R > values that have any practical meaning beyond E12 ohms or so. Most > practical insulation Rs may be around E12-E14 tops, unless you go to > sapphire. Up in that region, the R may be all within a material, or > include surface components like a film of dirt or moisture, or a > fingerprint. > > E11 resistors can be made to fairly high precision, and maybe E12 > nowadays. In the old days, higher values were made by stacking E11s - > like ten in series to get E12 with decent precision. The glass packaging > also limits how high it can go, due to leakage within and on the > surface. I once used a glass reed relay capsule as an ultra-high > resistance in a circuit. There was no precision or stability at all, but > it made a nice high resistor (probably E14-ish dry) even though there > was no element in there, and the circuit didn't care, as long as it was > very high, but not infinite. > > The specs on this HP unit are likely just the most extreme capability > taking maximum voltage over minimum current resolution, but any > measurements would tend to be very noisy and unstable anyway. Also, > testing at the extreme 1 kV makes the numbers seem more impressive, but > the voltage coefficient of resistance will pretty much be unpredictable. > > If this is a digital meter, then the other spec trick that tends to > obscure the real performance limit is that the ultimate resolution and > noise is that last digit - or even last two or three - that may may be > pretty jumpy, unless very long averaging time is used. > > There may be newer, fancier electrometers nowadays, but Keithley used to > be the standard for these in the old days, before several digits of DVM > resolution complicated the specs. They had a vibrating capacitor > electrometer with all-sapphire input structure back in the 1970s/80s I > think, that was the epitome of electrometers. I forget the model number, > but vaguely recall that it could reach the aA region full scale - not > that last digit of resolution thing. It's long obsolete, and I don't > think they ever made anything actually better - only added DVM digits to > less capable, conventional semiconductor amplifier techniques. If you > can find info on it, it's an interesting read. I found a pdf of the > manual years ago, but have no idea where it is now, or what info may > still be around. > > Ed > > ------------------------------ > > Message: 2 > Date: Fri, 2 Mar 2018 20:41:10 -0500 > From: "Mitch Van Ochten" <Mitch@vincentelectronics.com> > To: "'Discussion of precise voltage measurement'" <volt-nuts@febo.com> > Subject: Re: [volt-nuts] Precision high resistance measurements / > calibration of HP 4339B high-resistance meter. > Message-ID: <014e01d3b290$b54948f0$1fdbdad0$@vincentelectronics.com> > Content-Type: text/plain; charset="UTF-8" > > Here are specs from an older General Radio Bridge for observation. > > GR 1644-A specifications: > > Resistance Range: 1 kΩ to 1000 TΩ (10^3 to 10^15 Ω) in ten ranges. > Accuracy: 10^3 Ω to 10^10 Ω, ±1 %. After self-calibration: 10^10 to > 10^12 Ω, ±1%*; 10^13 Ω, ±2%; 10^14 Ω, ±10%; 10^15 Ω, ± one scale > division. > ΔR% Dial: ±5% range; accurate to ±0.2% or, for small changes, > to ±0.1%. > Test Voltage: Voltage accuracy is ±3% ±0.5 V. > Fixed Voltages** 10 20 50 100 200 500 1000 v > > Best regards, > > mitch > > -----Original Message----- > From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of ed breya > Sent: Friday, March 02, 2018 12:23 PM > To: volt-nuts@febo.com > Subject: Re: [volt-nuts] Precision high resistance measurements / calibration of HP 4339B high-resistance meter. > > Oops - I think I didn't send this message properly yesterday - here goes again. Ed > > Yes, David, unless you go to very extreme measures, you won't see real R values that have any practical meaning beyond E12 ohms or so. Most practical insulation Rs may be around E12-E14 tops, unless you go to sapphire. Up in that region, the R may be all within a material, or include surface components like a film of dirt or moisture, or a fingerprint. > > E11 resistors can be made to fairly high precision, and maybe E12 nowadays. In the old days, higher values were made by stacking E11s - like ten in series to get E12 with decent precision. The glass packaging also limits how high it can go, due to leakage within and on the surface. I once used a glass reed relay capsule as an ultra-high resistance in a circuit. There was no precision or stability at all, but it made a nice high resistor (probably E14-ish dry) even though there was no element in there, and the circuit didn't care, as long as it was very high, but not infinite. > > The specs on this HP unit are likely just the most extreme capability taking maximum voltage over minimum current resolution, but any measurements would tend to be very noisy and unstable anyway. Also, testing at the extreme 1 kV makes the numbers seem more impressive, but the voltage coefficient of resistance will pretty much be unpredictable. > > If this is a digital meter, then the other spec trick that tends to obscure the real performance limit is that the ultimate resolution and noise is that last digit - or even last two or three - that may may be pretty jumpy, unless very long averaging time is used. > > There may be newer, fancier electrometers nowadays, but Keithley used to be the standard for these in the old days, before several digits of DVM resolution complicated the specs. They had a vibrating capacitor electrometer with all-sapphire input structure back in the 1970s/80s I think, that was the epitome of electrometers. I forget the model number, but vaguely recall that it could reach the aA region full scale - not that last digit of resolution thing. It's long obsolete, and I don't think they ever made anything actually better - only added DVM digits to less capable, conventional semiconductor amplifier techniques. If you can find info on it, it's an interesting read. I found a pdf of the manual years ago, but have no idea where it is now, or what info may still be around. > > Ed > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.febo.com%2Fcgi-bin%2Fmailman%2Flistinfo%2Fvolt-nuts&data=02%7C01%7C%7Cb5818846d36e4ecd914908d581283c52%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C636556932161076932&sdata=cp01qd%2FkRw4yYSLYOqJ8RwUF8oOxtdqLVOmDal7gBhs%3D&reserved=0 https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.febo.com%2Fcgi-bin%2Fmailman%2Flistinfo%2Fvolt-nuts&data=02%7C01%7C%7Cb5818846d36e4ecd914908d581283c52%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C636556932161076932&sdata=cp01qd%2FkRw4yYSLYOqJ8RwUF8oOxtdqLVOmDal7gBhs%3D&reserved=0 > and follow the instructions there. > > ------------------------------ > > Subject: Digest Footer > > _______________________________________________ > volt-nuts mailing list > volt-nuts@febo.com > https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.febo.com%2Fcgi-bin%2Fmailman%2Flistinfo%2Fvolt-nuts&data=02%7C01%7C%7Cb5818846d36e4ecd914908d581283c52%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C636556932161076932&sdata=cp01qd%2FkRw4yYSLYOqJ8RwUF8oOxtdqLVOmDal7gBhs%3D&reserved=0 https://nam02.safelinks.protection.outlook.com/?url=https%3A%2F%2Fwww.febo.com%2Fcgi-bin%2Fmailman%2Flistinfo%2Fvolt-nuts&data=02%7C01%7C%7Cb5818846d36e4ecd914908d581283c52%7C84df9e7fe9f640afb435aaaaaaaaaaaa%7C1%7C0%7C636556932161076932&sdata=cp01qd%2FkRw4yYSLYOqJ8RwUF8oOxtdqLVOmDal7gBhs%3D&reserved=0 > > ------------------------------ > > End of volt-nuts Digest, Vol 103, Issue 3 > > ***************************************** > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >
EB
ed breya
Tue, Mar 6, 2018 11:37 PM

Yup - Keithley 640 - that must be the one. This stirred my memory
somewhat, and I just located info on the model 642 also, which was
apparently newer. The 642 went to (or back to) ultra-low bias MOSFETs,
while keeping sapphire insulation and a separate input head. The MOSFETs
need all kinds of offset, temperature, and bias current compensation.
The 642 also uses a few digits of DVM that obscure the real
capabilities, as I mentioned previously. The specs apparently show the
most sensitive range as 1 pA FS, so all stuff below E-12 A depends on
those digits to resolve. The manual recommends that extremely small
currents below 1 fA (the third digit down) be measured in charge mode. I
think this is to compensate for bias current and to average out some of
the 1/f noise.

The 640 on the other hand, can apparently reach 1 fA FS (1000x lower)
with 100 aA p-p (+/- 5%) noise on analog readout. Given a choice between
the two, I think I'd pick the 640, and hook a DVM to the output, and
average a whole bunch if necessary. I think the 640 uses a superior
front-end technology that maybe could be even further improved in the
middle and back end, while the 642 probably is as good as it can ever be
already.

Ed

Yup - Keithley 640 - that must be the one. This stirred my memory somewhat, and I just located info on the model 642 also, which was apparently newer. The 642 went to (or back to) ultra-low bias MOSFETs, while keeping sapphire insulation and a separate input head. The MOSFETs need all kinds of offset, temperature, and bias current compensation. The 642 also uses a few digits of DVM that obscure the real capabilities, as I mentioned previously. The specs apparently show the most sensitive range as 1 pA FS, so all stuff below E-12 A depends on those digits to resolve. The manual recommends that extremely small currents below 1 fA (the third digit down) be measured in charge mode. I think this is to compensate for bias current and to average out some of the 1/f noise. The 640 on the other hand, can apparently reach 1 fA FS (1000x lower) with 100 aA p-p (+/- 5%) noise on analog readout. Given a choice between the two, I think I'd pick the 640, and hook a DVM to the output, and average a whole bunch if necessary. I think the 640 uses a superior front-end technology that maybe could be even further improved in the middle and back end, while the 642 probably is as good as it can ever be already. Ed
EB
ed breya
Tue, Mar 6, 2018 11:52 PM

Another thing I noticed in these instruments - the highest R value used
is E12, even though decades higher would have been appropriate in
certain ranges. It shows that was about the practical limit for somewhat
decent precision and cost. Filling in the desired higher ranges had to
be done by adding complexity like more gain elsewhere, and writing the
specs accordingly.

Ed

Another thing I noticed in these instruments - the highest R value used is E12, even though decades higher would have been appropriate in certain ranges. It shows that was about the practical limit for somewhat decent precision and cost. Filling in the desired higher ranges had to be done by adding complexity like more gain elsewhere, and writing the specs accordingly. Ed
EB
ed breya
Thu, Mar 8, 2018 6:55 PM

Curiosity forced me to look around a little more for info on the 640.
The first link includes some info on the 640 with pictures. The
vibrating capacitor thing looks like a sort of vacuum tube. I thought
maybe it was custom made by or for Keithley, but it seems to actually
have been an off-shelf commercial product. The second link shows
pictures of what it likely is - XL7900, made by Philips back then. The
third link is to the data sheet that explains a lot.

https://doc.xdevs.com/doc/_Metrology/femto_ampere_current_source.pdf

http://lampes-et-tubes.info/sp/sp003.php?l=e

http://frank.pocnet.net/sheets/009/x/XL7900.pdf

I also found some info on the Varian/Cary 401 vibrating reed type
electrometer. It seems it was very common and used in a lot of
applications from the late 1960s on. From sparse technical info and
pictures, it appears to be actually built up from parts, into an
interesting machine. I could not find any manuals (without signing up
for some kind of subscription at document websites) or detailed info.

Ed

Curiosity forced me to look around a little more for info on the 640. The first link includes some info on the 640 with pictures. The vibrating capacitor thing looks like a sort of vacuum tube. I thought maybe it was custom made by or for Keithley, but it seems to actually have been an off-shelf commercial product. The second link shows pictures of what it likely is - XL7900, made by Philips back then. The third link is to the data sheet that explains a lot. https://doc.xdevs.com/doc/_Metrology/femto_ampere_current_source.pdf http://lampes-et-tubes.info/sp/sp003.php?l=e http://frank.pocnet.net/sheets/009/x/XL7900.pdf I also found some info on the Varian/Cary 401 vibrating reed type electrometer. It seems it was very common and used in a lot of applications from the late 1960s on. From sparse technical info and pictures, it appears to be actually built up from parts, into an interesting machine. I could not find any manuals (without signing up for some kind of subscription at document websites) or detailed info. Ed