Interpreting and Understanding Allen Deviation Results

C
CubeCentral
Wed, Nov 15, 2017 4:12 PM

Greetings, time-nuts!

After reading [ http://www.leapsecond.com/pages/adev/adev-why.htm ] I felt
that I better understood how an Allan Deviation is calculated and endeavored
to try an experiment.  It should be noted that I have a hobbyist-level
understanding of the concepts described and tools used below.  If my
thinking or test methodology is incorrect, please let me know so that I
might learn something.

A GPSDO with a 10MHz output was run into the EXT TIME BASE input on the back
of an HP5335A.
Then, the TIME BASE OUT on the back was run to the A input on the front of
the HP5335A.
My intention was to characterize the performance of the HP5335A counter
itself so that I might understand better future plots involving other GPSDO
and the counter's internal clock (which was bypassed for this test).

The settings of the HP5335A were as follows:
Gate Mode: Normal
Cycle: Normal

A Input ------------------------------
Trigger Adjust: Full left to 'Preset' detent
Z select  =  in  =  50ohm
x10 ATTN  =  in  =  x10 ATTN  (should have been out/off?)
Slope    =  out  =  up
AC        =  in  =  AC coupled
COMA      =  out  =  Not ComA
AutoTrig  =  out  =  Not Auto Tiggered (should have been in/on?)

(Tangentially, if someone has a good 'primer' or how-to resource detailing
Universal Counter operation, showing when/why/how to set the knobs in
certain situations it would be welcome!)

I then set the Time Lab V1.29 software to repeatedly acquire data for 12
hours, starting the next test as soon as I could.  This means that,
normally, a test was run during the day for 12 hours, and then overnight for
12 hours.

The results are shown here:  [ https://i.imgur.com/0sMVMfk.png ]  The
associated .TIM files are available upon request.

So, now we get to the heart of the matter and the questions this test and
results have raised.
I am trying to understand what the data is telling me about the test, and
therefore the character of the counter.

  1. Why are the plots a straight line from ~0.25s until ~100s?
  2. Why, after falling at the start, do the plots all seem to go back up
    from ~100s to ~1000s?
  3. What do the "peaks" mean, after the plot has fallen and begin to rise
    again?
  4. Why is the period from ~1000s to ~10000s so chaotic?
  5. The pattern "Fall to a minimum point, then rise to a peak, then fall
    again" seems to be prevalent.  What does that indicate?
  6. Why does that pattern in question (5) seem to repeat sometimes?  What is
    that showing me?

And finally, some general questions about looking at these plots.
a)  Would a "perfect" plot be a straight line falling from left to right?
(Meaning a hypothetical "ideal" source with perfect timing?)
b)  Is there some example showing plots from two different sources that then
describes why one source is better than the other (based upon the ADEV
plot)?
c)  I believe that if I understood the math better, these types of plots
would be more telling.  Without having to dive back into my college Calculus
or Statistics books, is there a good resource for me to be able to
understand this better?

Lastly, thank you for your patience and for keeping this brain-trust alive.
I am quite grateful for all the time and energy members pour into this list.
The archives have been a good source of learning material.

-Randal (at CubeCentral Labs...)
Greetings, time-nuts! After reading [ http://www.leapsecond.com/pages/adev/adev-why.htm ] I felt that I better understood how an Allan Deviation is calculated and endeavored to try an experiment. It should be noted that I have a hobbyist-level understanding of the concepts described and tools used below. If my thinking or test methodology is incorrect, please let me know so that I might learn something. A GPSDO with a 10MHz output was run into the EXT TIME BASE input on the back of an HP5335A. Then, the TIME BASE OUT on the back was run to the A input on the front of the HP5335A. My intention was to characterize the performance of the HP5335A counter itself so that I might understand better future plots involving other GPSDO and the counter's internal clock (which was bypassed for this test). The settings of the HP5335A were as follows: Gate Mode: Normal Cycle: Normal A Input ------------------------------ Trigger Adjust: Full left to 'Preset' detent Z select = in = 50ohm x10 ATTN = in = x10 ATTN (should have been out/off?) Slope = out = up AC = in = AC coupled COMA = out = Not ComA AutoTrig = out = Not Auto Tiggered (should have been in/on?) (Tangentially, if someone has a good 'primer' or how-to resource detailing Universal Counter operation, showing when/why/how to set the knobs in certain situations it would be welcome!) I then set the Time Lab V1.29 software to repeatedly acquire data for 12 hours, starting the next test as soon as I could. This means that, normally, a test was run during the day for 12 hours, and then overnight for 12 hours. The results are shown here: [ https://i.imgur.com/0sMVMfk.png ] The associated .TIM files are available upon request. So, now we get to the heart of the matter and the questions this test and results have raised. I am trying to understand what the data is telling me about the test, and therefore the character of the counter. 1) Why are the plots a straight line from ~0.25s until ~100s? 2) Why, after falling at the start, do the plots all seem to go back up from ~100s to ~1000s? 3) What do the "peaks" mean, after the plot has fallen and begin to rise again? 4) Why is the period from ~1000s to ~10000s so chaotic? 5) The pattern "Fall to a minimum point, then rise to a peak, then fall again" seems to be prevalent. What does that indicate? 6) Why does that pattern in question (5) seem to repeat sometimes? What is that showing me? And finally, some general questions about looking at these plots. a) Would a "perfect" plot be a straight line falling from left to right? (Meaning a hypothetical "ideal" source with perfect timing?) b) Is there some example showing plots from two different sources that then describes why one source is better than the other (based upon the ADEV plot)? c) I believe that if I understood the math better, these types of plots would be more telling. Without having to dive back into my college Calculus or Statistics books, is there a good resource for me to be able to understand this better? Lastly, thank you for your patience and for keeping this brain-trust alive. I am quite grateful for all the time and energy members pour into this list. The archives have been a good source of learning material. -Randal (at CubeCentral Labs...)
AK
Attila Kinali
Wed, Nov 15, 2017 11:46 PM

Hi,

There are other, more qualified people who should answer your questions,
but until they get around to it, let me give you some pointers to read
up upon.

First of all, you probably want to read [1] and [2]. Especially the
latter does explain the effects of the different noise types and
how they look like in the *DEV plots quite nicely.

On Wed, 15 Nov 2017 09:12:22 -0700
"CubeCentral" cubecentral@gmail.com wrote:

I then set the Time Lab V1.29 software to repeatedly acquire data for 12
hours, starting the next test as soon as I could.  This means that,
normally, a test was run during the day for 12 hours, and then overnight for
12 hours.

The results are shown here:  [ https://i.imgur.com/0sMVMfk.png ]  The
associated .TIM files are available upon request.

For this kind of test, I would rather use TDEV, as it shows more details
especially at longer taus, but if you are specifically looking for
frequency stability measures, ADEV (or MDEV) should be fine.

You should also switch on the error bars. Especially if you are looking
at the longer taus, close to your measurement length.

  1. Why are the plots a straight line from ~0.25s until ~100s?

Because they are dominated by white phase noise and flicker phase
noise which falls with 1/tau. The start point on the left side
is limited by your measurement precision, ie minimum resolution
of the instrument and its (white) noise.

  1. Why, after falling at the start, do the plots all seem to go back up
    from ~100s to ~1000s?

Because other noise types (white frequency, flicker frequency,...) become
dominant. What you see there is basically the instability of the measurement
electronics, the change in delays within the instrument due to temperature,
humidity, aging and other effects, trigger point changes, etc

  1. What do the "peaks" mean, after the plot has fallen and begin to rise
    again?

They are probably statistically insignificant, but it's impossible
to tell without the error bars. What wavy *DEV usually mean is,
that you have some periodic disturbance (A/C unit cycling, diurnal temperature
change, people walking in at specific times, train passing by on track nearby).

It's also a good idea to look at the phase plot or frequency plot
to see whether there are any deterministic effects that you can see.

  1. Why is the period from ~1000s to ~10000s so chaotic?

Statistical instability. You are measuring for 12h, that's 43200 seconds.
So at a tau of 10k, you only have roughly 4 samples. You cannot do any
valid statistics with that.

  1. The pattern "Fall to a minimum point, then rise to a peak, then fall
    again" seems to be prevalent.  What does that indicate?

Ideally, it should go down first (WPM, FPM), reach a minimum (WFM)
then go up again (FFM,...). But with such a short run it will not
be that way (statistical uncertainty). But then, very few plots
are really that way, because at the longer tau, the effects that
affect the measurement tend to be less and less gaus distributed
and you start to see patterns.

  1. Why does that pattern in question (5) seem to repeat sometimes?  What is
    that showing me?

The repetition is a prime characteristic of a periodic disturbance.
But in this case it's more likely that it's just the bad statistics
tricking you into thinking you see a pattern where there is none
(at least none of statistical significance).

And finally, some general questions about looking at these plots.
a)  Would a "perfect" plot be a straight line falling from left to right?
(Meaning a hypothetical "ideal" source with perfect timing?)

Yes. If you had only white and flicker phase noise. But this is
quite hard to achieve with frequency counters that have dead time
between measurements.

b)  Is there some example showing plots from two different sources that then
describes why one source is better than the other (based upon the ADEV
plot)?

Lower is better... but where it should be lower depends on your
target application.

c)  I believe that if I understood the math better, these types of plots
would be more telling.  Without having to dive back into my college Calculus
or Statistics books, is there a good resource for me to be able to
understand this better?

The two documents I listed below should get you started.
The problem is, if you really want to understand what's going
on you will need to understand what the different noise types
are, how they behave mathematically and that's where things
get... weird. If you go that road, you will soon learn that
your college calculus will only be a very rough guide, like a
1:25'0000 map while trying to navigate a city. While white
phase noise is relatively easy to understand with Probability 101,
the higher order noises are not.

I am currently trying to put a small document together of what
I've learned about noise, but it's increadibly hard as the math
is quite often beyond me. Not to mention the knowledge is very
fragmented over the fields of mathematics (statistics/probability,
fractals, stochastic calculus, fractional calculus, ...), physics
(solid state physics, fluid dynamics, ....) and engineering...
and each of these (sub-)fields using a different nomenclature
and different notation.

		Attila Kinali

[1] "Characterization of Clocks and Oscillators", NIST Technical Note 1337,
by Sulivan, Allan, Howe, Walls, 1990
http://tf.nist.gov/general/pdf/868.pdf

[2] "Handbook of Frequency Stability Analysis", NIST Special Publication 1065,
by Riley 2008
http://tf.nist.gov/timefreq/general/pdf/2220.pdf

PS: It's Allan Deviation, not Allen.

You know, the very powerful and the very stupid have one thing in common.
They don't alters their views to fit the facts, they alter the facts to
fit the views, which can be uncomfortable if you happen to be one of the
facts that needs altering.  -- The Doctor

Hi, There are other, more qualified people who should answer your questions, but until they get around to it, let me give you some pointers to read up upon. First of all, you probably want to read [1] and [2]. Especially the latter does explain the effects of the different noise types and how they look like in the *DEV plots quite nicely. On Wed, 15 Nov 2017 09:12:22 -0700 "CubeCentral" <cubecentral@gmail.com> wrote: > I then set the Time Lab V1.29 software to repeatedly acquire data for 12 > hours, starting the next test as soon as I could. This means that, > normally, a test was run during the day for 12 hours, and then overnight for > 12 hours. > The results are shown here: [ https://i.imgur.com/0sMVMfk.png ] The > associated .TIM files are available upon request. For this kind of test, I would rather use TDEV, as it shows more details especially at longer taus, but if you are specifically looking for frequency stability measures, ADEV (or MDEV) should be fine. You should also switch on the error bars. Especially if you are looking at the longer taus, close to your measurement length. > 1) Why are the plots a straight line from ~0.25s until ~100s? Because they are dominated by white phase noise and flicker phase noise which falls with 1/tau. The start point on the left side is limited by your measurement precision, ie minimum resolution of the instrument and its (white) noise. > 2) Why, after falling at the start, do the plots all seem to go back up > from ~100s to ~1000s? Because other noise types (white frequency, flicker frequency,...) become dominant. What you see there is basically the instability of the measurement electronics, the change in delays within the instrument due to temperature, humidity, aging and other effects, trigger point changes, etc > 3) What do the "peaks" mean, after the plot has fallen and begin to rise > again? They are probably statistically insignificant, but it's impossible to tell without the error bars. What wavy *DEV usually mean is, that you have some periodic disturbance (A/C unit cycling, diurnal temperature change, people walking in at specific times, train passing by on track nearby). It's also a good idea to look at the phase plot or frequency plot to see whether there are any deterministic effects that you can see. > 4) Why is the period from ~1000s to ~10000s so chaotic? Statistical instability. You are measuring for 12h, that's 43200 seconds. So at a tau of 10k, you only have roughly 4 samples. You cannot do any valid statistics with that. > 5) The pattern "Fall to a minimum point, then rise to a peak, then fall > again" seems to be prevalent. What does that indicate? Ideally, it should go down first (WPM, FPM), reach a minimum (WFM) then go up again (FFM,...). But with such a short run it will not be that way (statistical uncertainty). But then, very few plots are really that way, because at the longer tau, the effects that affect the measurement tend to be less and less gaus distributed and you start to see patterns. > 6) Why does that pattern in question (5) seem to repeat sometimes? What is > that showing me? The repetition is a prime characteristic of a periodic disturbance. But in this case it's more likely that it's just the bad statistics tricking you into thinking you see a pattern where there is none (at least none of statistical significance). > And finally, some general questions about looking at these plots. > a) Would a "perfect" plot be a straight line falling from left to right? > (Meaning a hypothetical "ideal" source with perfect timing?) Yes. If you had only white and flicker phase noise. But this is quite hard to achieve with frequency counters that have dead time between measurements. > b) Is there some example showing plots from two different sources that then > describes why one source is better than the other (based upon the ADEV > plot)? Lower is better... but where it should be lower depends on your target application. > c) I believe that if I understood the math better, these types of plots > would be more telling. Without having to dive back into my college Calculus > or Statistics books, is there a good resource for me to be able to > understand this better? The two documents I listed below should get you started. The problem is, if you really want to understand what's going on you will need to understand what the different noise types are, how they behave mathematically and that's where things get... weird. If you go that road, you will soon learn that your college calculus will only be a very rough guide, like a 1:25'0000 map while trying to navigate a city. While white phase noise is relatively easy to understand with Probability 101, the higher order noises are not. I am currently trying to put a small document together of what I've learned about noise, but it's increadibly hard as the math is quite often beyond me. Not to mention the knowledge is very fragmented over the fields of mathematics (statistics/probability, fractals, stochastic calculus, fractional calculus, ...), physics (solid state physics, fluid dynamics, ....) and engineering... and each of these (sub-)fields using a different nomenclature and different notation. Attila Kinali [1] "Characterization of Clocks and Oscillators", NIST Technical Note 1337, by Sulivan, Allan, Howe, Walls, 1990 http://tf.nist.gov/general/pdf/868.pdf [2] "Handbook of Frequency Stability Analysis", NIST Special Publication 1065, by Riley 2008 http://tf.nist.gov/timefreq/general/pdf/2220.pdf PS: It's Allan Deviation, not Allen. -- You know, the very powerful and the very stupid have one thing in common. They don't alters their views to fit the facts, they alter the facts to fit the views, which can be uncomfortable if you happen to be one of the facts that needs altering. -- The Doctor
MG
Mike Garvey
Thu, Nov 16, 2017 2:59 AM

Could you post some phase plots?  The data you show is not 1/tau and very likely not white phase noise.
Mike

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of CubeCentral
Sent: Wednesday, November 15, 2017 11:12
To: time-nuts@febo.com
Subject: [time-nuts] Interpreting and Understanding Allen Deviation Results

Greetings, time-nuts!

After reading [ http://www.leapsecond.com/pages/adev/adev-why.htm ] I felt that I better understood how an Allan Deviation is calculated and endeavored to try an experiment.  It should be noted that I have a hobbyist-level understanding of the concepts described and tools used below.  If my thinking or test methodology is incorrect, please let me know so that I might learn something.

A GPSDO with a 10MHz output was run into the EXT TIME BASE input on the back of an HP5335A.
Then, the TIME BASE OUT on the back was run to the A input on the front of the HP5335A.
My intention was to characterize the performance of the HP5335A counter itself so that I might understand better future plots involving other GPSDO and the counter's internal clock (which was bypassed for this test).

The settings of the HP5335A were as follows:
Gate Mode: Normal
Cycle: Normal

A Input ------------------------------
Trigger Adjust: Full left to 'Preset' detent
Z select  =  in  =  50ohm
x10 ATTN  =  in  =  x10 ATTN  (should have been out/off?)
Slope    =  out  =  up
AC        =  in  =  AC coupled
COMA      =  out  =  Not ComA
AutoTrig  =  out  =  Not Auto Tiggered (should have been in/on?)

(Tangentially, if someone has a good 'primer' or how-to resource detailing Universal Counter operation, showing when/why/how to set the knobs in certain situations it would be welcome!)

I then set the Time Lab V1.29 software to repeatedly acquire data for 12 hours, starting the next test as soon as I could.  This means that, normally, a test was run during the day for 12 hours, and then overnight for
12 hours.

The results are shown here:  [ https://i.imgur.com/0sMVMfk.png ]  The associated .TIM files are available upon request.

So, now we get to the heart of the matter and the questions this test and results have raised.
I am trying to understand what the data is telling me about the test, and therefore the character of the counter.

  1. Why are the plots a straight line from ~0.25s until ~100s?
  2. Why, after falling at the start, do the plots all seem to go back up from ~100s to ~1000s?
  3. What do the "peaks" mean, after the plot has fallen and begin to rise again?
  4. Why is the period from ~1000s to ~10000s so chaotic?
  5. The pattern "Fall to a minimum point, then rise to a peak, then fall again" seems to be prevalent.  What does that indicate?
  6. Why does that pattern in question (5) seem to repeat sometimes?  What is that showing me?

And finally, some general questions about looking at these plots.
a)  Would a "perfect" plot be a straight line falling from left to right?
(Meaning a hypothetical "ideal" source with perfect timing?)
b)  Is there some example showing plots from two different sources that then describes why one source is better than the other (based upon the ADEV plot)?
c)  I believe that if I understood the math better, these types of plots would be more telling.  Without having to dive back into my college Calculus or Statistics books, is there a good resource for me to be able to understand this better?

Lastly, thank you for your patience and for keeping this brain-trust alive.
I am quite grateful for all the time and energy members pour into this list.
The archives have been a good source of learning material.

-Randal (at CubeCentral Labs...)

time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Could you post some phase plots? The data you show is not 1/tau and very likely not white phase noise. Mike -----Original Message----- From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of CubeCentral Sent: Wednesday, November 15, 2017 11:12 To: time-nuts@febo.com Subject: [time-nuts] Interpreting and Understanding Allen Deviation Results Greetings, time-nuts! After reading [ http://www.leapsecond.com/pages/adev/adev-why.htm ] I felt that I better understood how an Allan Deviation is calculated and endeavored to try an experiment. It should be noted that I have a hobbyist-level understanding of the concepts described and tools used below. If my thinking or test methodology is incorrect, please let me know so that I might learn something. A GPSDO with a 10MHz output was run into the EXT TIME BASE input on the back of an HP5335A. Then, the TIME BASE OUT on the back was run to the A input on the front of the HP5335A. My intention was to characterize the performance of the HP5335A counter itself so that I might understand better future plots involving other GPSDO and the counter's internal clock (which was bypassed for this test). The settings of the HP5335A were as follows: Gate Mode: Normal Cycle: Normal A Input ------------------------------ Trigger Adjust: Full left to 'Preset' detent Z select = in = 50ohm x10 ATTN = in = x10 ATTN (should have been out/off?) Slope = out = up AC = in = AC coupled COMA = out = Not ComA AutoTrig = out = Not Auto Tiggered (should have been in/on?) (Tangentially, if someone has a good 'primer' or how-to resource detailing Universal Counter operation, showing when/why/how to set the knobs in certain situations it would be welcome!) I then set the Time Lab V1.29 software to repeatedly acquire data for 12 hours, starting the next test as soon as I could. This means that, normally, a test was run during the day for 12 hours, and then overnight for 12 hours. The results are shown here: [ https://i.imgur.com/0sMVMfk.png ] The associated .TIM files are available upon request. So, now we get to the heart of the matter and the questions this test and results have raised. I am trying to understand what the data is telling me about the test, and therefore the character of the counter. 1) Why are the plots a straight line from ~0.25s until ~100s? 2) Why, after falling at the start, do the plots all seem to go back up from ~100s to ~1000s? 3) What do the "peaks" mean, after the plot has fallen and begin to rise again? 4) Why is the period from ~1000s to ~10000s so chaotic? 5) The pattern "Fall to a minimum point, then rise to a peak, then fall again" seems to be prevalent. What does that indicate? 6) Why does that pattern in question (5) seem to repeat sometimes? What is that showing me? And finally, some general questions about looking at these plots. a) Would a "perfect" plot be a straight line falling from left to right? (Meaning a hypothetical "ideal" source with perfect timing?) b) Is there some example showing plots from two different sources that then describes why one source is better than the other (based upon the ADEV plot)? c) I believe that if I understood the math better, these types of plots would be more telling. Without having to dive back into my college Calculus or Statistics books, is there a good resource for me to be able to understand this better? Lastly, thank you for your patience and for keeping this brain-trust alive. I am quite grateful for all the time and energy members pour into this list. The archives have been a good source of learning material. -Randal (at CubeCentral Labs...) _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
AB
Azelio Boriani
Thu, Nov 16, 2017 9:10 AM

As already stated here, the best measurement mode is the time-interval
mode. The 5335A is a 2ns single-shot resolution counter. Use the PPS
output from the GPSDO, route it to the A (start) input and to a
coaxial cable used as a delay line (10m, 50ns, should be enough). The
other end of the cable into the B input (stop), select the time
interval mode TIME A -> B. Let the internal reference clock the
counter. Set trigger levels and the various parameter to get stable
readings and collect your data.

On Thu, Nov 16, 2017 at 3:59 AM, Mike Garvey r3m1g4@verizon.net wrote:

Could you post some phase plots?  The data you show is not 1/tau and very likely not white phase noise.
Mike

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of CubeCentral
Sent: Wednesday, November 15, 2017 11:12
To: time-nuts@febo.com
Subject: [time-nuts] Interpreting and Understanding Allen Deviation Results

Greetings, time-nuts!

After reading [ http://www.leapsecond.com/pages/adev/adev-why.htm ] I felt that I better understood how an Allan Deviation is calculated and endeavored to try an experiment.  It should be noted that I have a hobbyist-level understanding of the concepts described and tools used below.  If my thinking or test methodology is incorrect, please let me know so that I might learn something.

A GPSDO with a 10MHz output was run into the EXT TIME BASE input on the back of an HP5335A.
Then, the TIME BASE OUT on the back was run to the A input on the front of the HP5335A.
My intention was to characterize the performance of the HP5335A counter itself so that I might understand better future plots involving other GPSDO and the counter's internal clock (which was bypassed for this test).

The settings of the HP5335A were as follows:
Gate Mode: Normal
Cycle: Normal

A Input ------------------------------
Trigger Adjust: Full left to 'Preset' detent
Z select  =  in  =  50ohm
x10 ATTN  =  in  =  x10 ATTN  (should have been out/off?)
Slope    =  out  =  up
AC        =  in  =  AC coupled
COMA      =  out  =  Not ComA
AutoTrig  =  out  =  Not Auto Tiggered (should have been in/on?)

(Tangentially, if someone has a good 'primer' or how-to resource detailing Universal Counter operation, showing when/why/how to set the knobs in certain situations it would be welcome!)

I then set the Time Lab V1.29 software to repeatedly acquire data for 12 hours, starting the next test as soon as I could.  This means that, normally, a test was run during the day for 12 hours, and then overnight for
12 hours.

The results are shown here:  [ https://i.imgur.com/0sMVMfk.png ]  The associated .TIM files are available upon request.

So, now we get to the heart of the matter and the questions this test and results have raised.
I am trying to understand what the data is telling me about the test, and therefore the character of the counter.

  1. Why are the plots a straight line from ~0.25s until ~100s?
  2. Why, after falling at the start, do the plots all seem to go back up from ~100s to ~1000s?
  3. What do the "peaks" mean, after the plot has fallen and begin to rise again?
  4. Why is the period from ~1000s to ~10000s so chaotic?
  5. The pattern "Fall to a minimum point, then rise to a peak, then fall again" seems to be prevalent.  What does that indicate?
  6. Why does that pattern in question (5) seem to repeat sometimes?  What is that showing me?

And finally, some general questions about looking at these plots.
a)  Would a "perfect" plot be a straight line falling from left to right?
(Meaning a hypothetical "ideal" source with perfect timing?)
b)  Is there some example showing plots from two different sources that then describes why one source is better than the other (based upon the ADEV plot)?
c)  I believe that if I understood the math better, these types of plots would be more telling.  Without having to dive back into my college Calculus or Statistics books, is there a good resource for me to be able to understand this better?

Lastly, thank you for your patience and for keeping this brain-trust alive.
I am quite grateful for all the time and energy members pour into this list.
The archives have been a good source of learning material.

     -Randal (at CubeCentral Labs...)

time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

As already stated here, the best measurement mode is the time-interval mode. The 5335A is a 2ns single-shot resolution counter. Use the PPS output from the GPSDO, route it to the A (start) input and to a coaxial cable used as a delay line (10m, 50ns, should be enough). The other end of the cable into the B input (stop), select the time interval mode TIME A -> B. Let the internal reference clock the counter. Set trigger levels and the various parameter to get stable readings and collect your data. On Thu, Nov 16, 2017 at 3:59 AM, Mike Garvey <r3m1g4@verizon.net> wrote: > Could you post some phase plots? The data you show is not 1/tau and very likely not white phase noise. > Mike > > -----Original Message----- > From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of CubeCentral > Sent: Wednesday, November 15, 2017 11:12 > To: time-nuts@febo.com > Subject: [time-nuts] Interpreting and Understanding Allen Deviation Results > > Greetings, time-nuts! > > After reading [ http://www.leapsecond.com/pages/adev/adev-why.htm ] I felt that I better understood how an Allan Deviation is calculated and endeavored to try an experiment. It should be noted that I have a hobbyist-level understanding of the concepts described and tools used below. If my thinking or test methodology is incorrect, please let me know so that I might learn something. > > A GPSDO with a 10MHz output was run into the EXT TIME BASE input on the back of an HP5335A. > Then, the TIME BASE OUT on the back was run to the A input on the front of the HP5335A. > My intention was to characterize the performance of the HP5335A counter itself so that I might understand better future plots involving other GPSDO and the counter's internal clock (which was bypassed for this test). > > The settings of the HP5335A were as follows: > Gate Mode: Normal > Cycle: Normal > > A Input ------------------------------ > Trigger Adjust: Full left to 'Preset' detent > Z select = in = 50ohm > x10 ATTN = in = x10 ATTN (should have been out/off?) > Slope = out = up > AC = in = AC coupled > COMA = out = Not ComA > AutoTrig = out = Not Auto Tiggered (should have been in/on?) > > (Tangentially, if someone has a good 'primer' or how-to resource detailing Universal Counter operation, showing when/why/how to set the knobs in certain situations it would be welcome!) > > I then set the Time Lab V1.29 software to repeatedly acquire data for 12 hours, starting the next test as soon as I could. This means that, normally, a test was run during the day for 12 hours, and then overnight for > 12 hours. > > The results are shown here: [ https://i.imgur.com/0sMVMfk.png ] The associated .TIM files are available upon request. > > So, now we get to the heart of the matter and the questions this test and results have raised. > I am trying to understand what the data is telling me about the test, and therefore the character of the counter. > > 1) Why are the plots a straight line from ~0.25s until ~100s? > 2) Why, after falling at the start, do the plots all seem to go back up from ~100s to ~1000s? > 3) What do the "peaks" mean, after the plot has fallen and begin to rise again? > 4) Why is the period from ~1000s to ~10000s so chaotic? > 5) The pattern "Fall to a minimum point, then rise to a peak, then fall again" seems to be prevalent. What does that indicate? > 6) Why does that pattern in question (5) seem to repeat sometimes? What is that showing me? > > And finally, some general questions about looking at these plots. > a) Would a "perfect" plot be a straight line falling from left to right? > (Meaning a hypothetical "ideal" source with perfect timing?) > b) Is there some example showing plots from two different sources that then describes why one source is better than the other (based upon the ADEV plot)? > c) I believe that if I understood the math better, these types of plots would be more telling. Without having to dive back into my college Calculus or Statistics books, is there a good resource for me to be able to understand this better? > > Lastly, thank you for your patience and for keeping this brain-trust alive. > I am quite grateful for all the time and energy members pour into this list. > The archives have been a good source of learning material. > > -Randal (at CubeCentral Labs...) > > > > > > > > > > > > > > > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
BK
Bob kb8tq
Thu, Nov 16, 2017 2:57 PM

Hi

One “cute” thing to do when looking at GPSDO’s or GPS modules is to use the
“cable delay” setting. It will allow you to move the pps of one unit relative to the
pps of the other one. You then can be sure of which pps happens first. That makes
the A to B measurement much easier to analyze.

Short intervals also can lessen the impact of the time base accuracy in the counter
( you always are measuring a microsecond or so to a nanosecond resolution). Indeed
there are other issues (like jitter) that still are an issue.

Bob

On Nov 16, 2017, at 4:10 AM, Azelio Boriani azelio.boriani@gmail.com wrote:

As already stated here, the best measurement mode is the time-interval
mode. The 5335A is a 2ns single-shot resolution counter. Use the PPS
output from the GPSDO, route it to the A (start) input and to a
coaxial cable used as a delay line (10m, 50ns, should be enough). The
other end of the cable into the B input (stop), select the time
interval mode TIME A -> B. Let the internal reference clock the
counter. Set trigger levels and the various parameter to get stable
readings and collect your data.

On Thu, Nov 16, 2017 at 3:59 AM, Mike Garvey r3m1g4@verizon.net wrote:

Could you post some phase plots?  The data you show is not 1/tau and very likely not white phase noise.
Mike

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of CubeCentral
Sent: Wednesday, November 15, 2017 11:12
To: time-nuts@febo.com
Subject: [time-nuts] Interpreting and Understanding Allen Deviation Results

Greetings, time-nuts!

After reading [ http://www.leapsecond.com/pages/adev/adev-why.htm ] I felt that I better understood how an Allan Deviation is calculated and endeavored to try an experiment.  It should be noted that I have a hobbyist-level understanding of the concepts described and tools used below.  If my thinking or test methodology is incorrect, please let me know so that I might learn something.

A GPSDO with a 10MHz output was run into the EXT TIME BASE input on the back of an HP5335A.
Then, the TIME BASE OUT on the back was run to the A input on the front of the HP5335A.
My intention was to characterize the performance of the HP5335A counter itself so that I might understand better future plots involving other GPSDO and the counter's internal clock (which was bypassed for this test).

The settings of the HP5335A were as follows:
Gate Mode: Normal
Cycle: Normal

A Input ------------------------------
Trigger Adjust: Full left to 'Preset' detent
Z select  =  in  =  50ohm
x10 ATTN  =  in  =  x10 ATTN  (should have been out/off?)
Slope    =  out  =  up
AC        =  in  =  AC coupled
COMA      =  out  =  Not ComA
AutoTrig  =  out  =  Not Auto Tiggered (should have been in/on?)

(Tangentially, if someone has a good 'primer' or how-to resource detailing Universal Counter operation, showing when/why/how to set the knobs in certain situations it would be welcome!)

I then set the Time Lab V1.29 software to repeatedly acquire data for 12 hours, starting the next test as soon as I could.  This means that, normally, a test was run during the day for 12 hours, and then overnight for
12 hours.

The results are shown here:  [ https://i.imgur.com/0sMVMfk.png ]  The associated .TIM files are available upon request.

So, now we get to the heart of the matter and the questions this test and results have raised.
I am trying to understand what the data is telling me about the test, and therefore the character of the counter.

  1. Why are the plots a straight line from ~0.25s until ~100s?
  2. Why, after falling at the start, do the plots all seem to go back up from ~100s to ~1000s?
  3. What do the "peaks" mean, after the plot has fallen and begin to rise again?
  4. Why is the period from ~1000s to ~10000s so chaotic?
  5. The pattern "Fall to a minimum point, then rise to a peak, then fall again" seems to be prevalent.  What does that indicate?
  6. Why does that pattern in question (5) seem to repeat sometimes?  What is that showing me?

And finally, some general questions about looking at these plots.
a)  Would a "perfect" plot be a straight line falling from left to right?
(Meaning a hypothetical "ideal" source with perfect timing?)
b)  Is there some example showing plots from two different sources that then describes why one source is better than the other (based upon the ADEV plot)?
c)  I believe that if I understood the math better, these types of plots would be more telling.  Without having to dive back into my college Calculus or Statistics books, is there a good resource for me to be able to understand this better?

Lastly, thank you for your patience and for keeping this brain-trust alive.
I am quite grateful for all the time and energy members pour into this list.
The archives have been a good source of learning material.

    -Randal (at CubeCentral Labs...)

time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi One “cute” thing to do when looking at GPSDO’s or GPS modules is to use the “cable delay” setting. It will allow you to move the pps of one unit relative to the pps of the other one. You then can be sure of which pps happens first. That makes the A to B measurement much easier to analyze. Short intervals also can lessen the impact of the time base accuracy in the counter ( you always are measuring a microsecond or so to a nanosecond resolution). Indeed there are other issues (like jitter) that still are an issue. Bob > On Nov 16, 2017, at 4:10 AM, Azelio Boriani <azelio.boriani@gmail.com> wrote: > > As already stated here, the best measurement mode is the time-interval > mode. The 5335A is a 2ns single-shot resolution counter. Use the PPS > output from the GPSDO, route it to the A (start) input and to a > coaxial cable used as a delay line (10m, 50ns, should be enough). The > other end of the cable into the B input (stop), select the time > interval mode TIME A -> B. Let the internal reference clock the > counter. Set trigger levels and the various parameter to get stable > readings and collect your data. > > On Thu, Nov 16, 2017 at 3:59 AM, Mike Garvey <r3m1g4@verizon.net> wrote: >> Could you post some phase plots? The data you show is not 1/tau and very likely not white phase noise. >> Mike >> >> -----Original Message----- >> From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of CubeCentral >> Sent: Wednesday, November 15, 2017 11:12 >> To: time-nuts@febo.com >> Subject: [time-nuts] Interpreting and Understanding Allen Deviation Results >> >> Greetings, time-nuts! >> >> After reading [ http://www.leapsecond.com/pages/adev/adev-why.htm ] I felt that I better understood how an Allan Deviation is calculated and endeavored to try an experiment. It should be noted that I have a hobbyist-level understanding of the concepts described and tools used below. If my thinking or test methodology is incorrect, please let me know so that I might learn something. >> >> A GPSDO with a 10MHz output was run into the EXT TIME BASE input on the back of an HP5335A. >> Then, the TIME BASE OUT on the back was run to the A input on the front of the HP5335A. >> My intention was to characterize the performance of the HP5335A counter itself so that I might understand better future plots involving other GPSDO and the counter's internal clock (which was bypassed for this test). >> >> The settings of the HP5335A were as follows: >> Gate Mode: Normal >> Cycle: Normal >> >> A Input ------------------------------ >> Trigger Adjust: Full left to 'Preset' detent >> Z select = in = 50ohm >> x10 ATTN = in = x10 ATTN (should have been out/off?) >> Slope = out = up >> AC = in = AC coupled >> COMA = out = Not ComA >> AutoTrig = out = Not Auto Tiggered (should have been in/on?) >> >> (Tangentially, if someone has a good 'primer' or how-to resource detailing Universal Counter operation, showing when/why/how to set the knobs in certain situations it would be welcome!) >> >> I then set the Time Lab V1.29 software to repeatedly acquire data for 12 hours, starting the next test as soon as I could. This means that, normally, a test was run during the day for 12 hours, and then overnight for >> 12 hours. >> >> The results are shown here: [ https://i.imgur.com/0sMVMfk.png ] The associated .TIM files are available upon request. >> >> So, now we get to the heart of the matter and the questions this test and results have raised. >> I am trying to understand what the data is telling me about the test, and therefore the character of the counter. >> >> 1) Why are the plots a straight line from ~0.25s until ~100s? >> 2) Why, after falling at the start, do the plots all seem to go back up from ~100s to ~1000s? >> 3) What do the "peaks" mean, after the plot has fallen and begin to rise again? >> 4) Why is the period from ~1000s to ~10000s so chaotic? >> 5) The pattern "Fall to a minimum point, then rise to a peak, then fall again" seems to be prevalent. What does that indicate? >> 6) Why does that pattern in question (5) seem to repeat sometimes? What is that showing me? >> >> And finally, some general questions about looking at these plots. >> a) Would a "perfect" plot be a straight line falling from left to right? >> (Meaning a hypothetical "ideal" source with perfect timing?) >> b) Is there some example showing plots from two different sources that then describes why one source is better than the other (based upon the ADEV plot)? >> c) I believe that if I understood the math better, these types of plots would be more telling. Without having to dive back into my college Calculus or Statistics books, is there a good resource for me to be able to understand this better? >> >> Lastly, thank you for your patience and for keeping this brain-trust alive. >> I am quite grateful for all the time and energy members pour into this list. >> The archives have been a good source of learning material. >> >> -Randal (at CubeCentral Labs...) >> >> >> >> >> >> >> >> >> >> >> >> >> >> >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
J
Jerry
Thu, Nov 16, 2017 3:37 PM

Bob,

Do you mean then you do not need to put a physical long length of cable for the delay, just do it in software or do you do both?

Jerry, NY2KW

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob kb8tq
Sent: Thursday, November 16, 2017 9:58 AM
To: Discussion of precise time and frequency measurement time-nuts@febo.com
Subject: Re: [time-nuts] Interpreting and Understanding Allen Deviation Results

Hi

One “cute” thing to do when looking at GPSDO’s or GPS modules is to use the “cable delay” setting. It will allow you to move the pps of one unit relative to the pps of the other one. You then can be sure of which pps happens first. That makes the A to B measurement much easier to analyze.

Short intervals also can lessen the impact of the time base accuracy in the counter ( you always are measuring a microsecond or so to a nanosecond resolution). Indeed there are other issues (like jitter) that still are an issue.

Bob

On Nov 16, 2017, at 4:10 AM, Azelio Boriani azelio.boriani@gmail.com wrote:

As already stated here, the best measurement mode is the time-interval
mode. The 5335A is a 2ns single-shot resolution counter. Use the PPS
output from the GPSDO, route it to the A (start) input and to a
coaxial cable used as a delay line (10m, 50ns, should be enough). The
other end of the cable into the B input (stop), select the time
interval mode TIME A -> B. Let the internal reference clock the
counter. Set trigger levels and the various parameter to get stable
readings and collect your data.

On Thu, Nov 16, 2017 at 3:59 AM, Mike Garvey r3m1g4@verizon.net wrote:

Could you post some phase plots?  The data you show is not 1/tau and very likely not white phase noise.
Mike

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of
CubeCentral
Sent: Wednesday, November 15, 2017 11:12
To: time-nuts@febo.com
Subject: [time-nuts] Interpreting and Understanding Allen Deviation
Results

Greetings, time-nuts!

After reading [ http://www.leapsecond.com/pages/adev/adev-why.htm ] I felt that I better understood how an Allan Deviation is calculated and endeavored to try an experiment.  It should be noted that I have a hobbyist-level understanding of the concepts described and tools used below.  If my thinking or test methodology is incorrect, please let me know so that I might learn something.

A GPSDO with a 10MHz output was run into the EXT TIME BASE input on the back of an HP5335A.
Then, the TIME BASE OUT on the back was run to the A input on the front of the HP5335A.
My intention was to characterize the performance of the HP5335A counter itself so that I might understand better future plots involving other GPSDO and the counter's internal clock (which was bypassed for this test).

The settings of the HP5335A were as follows:
Gate Mode: Normal
Cycle: Normal

A Input ------------------------------ Trigger Adjust: Full left to
'Preset' detent
Z select  =  in  =  50ohm
x10 ATTN  =  in  =  x10 ATTN  (should have been out/off?)
Slope    =  out  =  up
AC        =  in  =  AC coupled
COMA      =  out  =  Not ComA
AutoTrig  =  out  =  Not Auto Tiggered (should have been in/on?)

(Tangentially, if someone has a good 'primer' or how-to resource
detailing Universal Counter operation, showing when/why/how to set
the knobs in certain situations it would be welcome!)

I then set the Time Lab V1.29 software to repeatedly acquire data for
12 hours, starting the next test as soon as I could.  This means
that, normally, a test was run during the day for 12 hours, and then
overnight for
12 hours.

The results are shown here:  [ https://i.imgur.com/0sMVMfk.png ]  The associated .TIM files are available upon request.

So, now we get to the heart of the matter and the questions this test and results have raised.
I am trying to understand what the data is telling me about the test, and therefore the character of the counter.

  1. Why are the plots a straight line from ~0.25s until ~100s?
  2. Why, after falling at the start, do the plots all seem to go back up from ~100s to ~1000s?
  3. What do the "peaks" mean, after the plot has fallen and begin to rise again?
  4. Why is the period from ~1000s to ~10000s so chaotic?
  5. The pattern "Fall to a minimum point, then rise to a peak, then fall again" seems to be prevalent.  What does that indicate?
  6. Why does that pattern in question (5) seem to repeat sometimes?  What is that showing me?

And finally, some general questions about looking at these plots.
a)  Would a "perfect" plot be a straight line falling from left to right?
(Meaning a hypothetical "ideal" source with perfect timing?)
b)  Is there some example showing plots from two different sources that then describes why one source is better than the other (based upon the ADEV plot)?
c)  I believe that if I understood the math better, these types of plots would be more telling.  Without having to dive back into my college Calculus or Statistics books, is there a good resource for me to be able to understand this better?

Lastly, thank you for your patience and for keeping this brain-trust alive.
I am quite grateful for all the time and energy members pour into this list.
The archives have been a good source of learning material.

    -Randal (at CubeCentral Labs...)

time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Bob, Do you mean then you do not need to put a physical long length of cable for the delay, just do it in software or do you do both? Jerry, NY2KW -----Original Message----- From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob kb8tq Sent: Thursday, November 16, 2017 9:58 AM To: Discussion of precise time and frequency measurement <time-nuts@febo.com> Subject: Re: [time-nuts] Interpreting and Understanding Allen Deviation Results Hi One “cute” thing to do when looking at GPSDO’s or GPS modules is to use the “cable delay” setting. It will allow you to move the pps of one unit relative to the pps of the other one. You then can be sure of which pps happens first. That makes the A to B measurement much easier to analyze. Short intervals also can lessen the impact of the time base accuracy in the counter ( you always are measuring a microsecond or so to a nanosecond resolution). Indeed there are other issues (like jitter) that still are an issue. Bob > On Nov 16, 2017, at 4:10 AM, Azelio Boriani <azelio.boriani@gmail.com> wrote: > > As already stated here, the best measurement mode is the time-interval > mode. The 5335A is a 2ns single-shot resolution counter. Use the PPS > output from the GPSDO, route it to the A (start) input and to a > coaxial cable used as a delay line (10m, 50ns, should be enough). The > other end of the cable into the B input (stop), select the time > interval mode TIME A -> B. Let the internal reference clock the > counter. Set trigger levels and the various parameter to get stable > readings and collect your data. > > On Thu, Nov 16, 2017 at 3:59 AM, Mike Garvey <r3m1g4@verizon.net> wrote: >> Could you post some phase plots? The data you show is not 1/tau and very likely not white phase noise. >> Mike >> >> -----Original Message----- >> From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of >> CubeCentral >> Sent: Wednesday, November 15, 2017 11:12 >> To: time-nuts@febo.com >> Subject: [time-nuts] Interpreting and Understanding Allen Deviation >> Results >> >> Greetings, time-nuts! >> >> After reading [ http://www.leapsecond.com/pages/adev/adev-why.htm ] I felt that I better understood how an Allan Deviation is calculated and endeavored to try an experiment. It should be noted that I have a hobbyist-level understanding of the concepts described and tools used below. If my thinking or test methodology is incorrect, please let me know so that I might learn something. >> >> A GPSDO with a 10MHz output was run into the EXT TIME BASE input on the back of an HP5335A. >> Then, the TIME BASE OUT on the back was run to the A input on the front of the HP5335A. >> My intention was to characterize the performance of the HP5335A counter itself so that I might understand better future plots involving other GPSDO and the counter's internal clock (which was bypassed for this test). >> >> The settings of the HP5335A were as follows: >> Gate Mode: Normal >> Cycle: Normal >> >> A Input ------------------------------ Trigger Adjust: Full left to >> 'Preset' detent >> Z select = in = 50ohm >> x10 ATTN = in = x10 ATTN (should have been out/off?) >> Slope = out = up >> AC = in = AC coupled >> COMA = out = Not ComA >> AutoTrig = out = Not Auto Tiggered (should have been in/on?) >> >> (Tangentially, if someone has a good 'primer' or how-to resource >> detailing Universal Counter operation, showing when/why/how to set >> the knobs in certain situations it would be welcome!) >> >> I then set the Time Lab V1.29 software to repeatedly acquire data for >> 12 hours, starting the next test as soon as I could. This means >> that, normally, a test was run during the day for 12 hours, and then >> overnight for >> 12 hours. >> >> The results are shown here: [ https://i.imgur.com/0sMVMfk.png ] The associated .TIM files are available upon request. >> >> So, now we get to the heart of the matter and the questions this test and results have raised. >> I am trying to understand what the data is telling me about the test, and therefore the character of the counter. >> >> 1) Why are the plots a straight line from ~0.25s until ~100s? >> 2) Why, after falling at the start, do the plots all seem to go back up from ~100s to ~1000s? >> 3) What do the "peaks" mean, after the plot has fallen and begin to rise again? >> 4) Why is the period from ~1000s to ~10000s so chaotic? >> 5) The pattern "Fall to a minimum point, then rise to a peak, then fall again" seems to be prevalent. What does that indicate? >> 6) Why does that pattern in question (5) seem to repeat sometimes? What is that showing me? >> >> And finally, some general questions about looking at these plots. >> a) Would a "perfect" plot be a straight line falling from left to right? >> (Meaning a hypothetical "ideal" source with perfect timing?) >> b) Is there some example showing plots from two different sources that then describes why one source is better than the other (based upon the ADEV plot)? >> c) I believe that if I understood the math better, these types of plots would be more telling. Without having to dive back into my college Calculus or Statistics books, is there a good resource for me to be able to understand this better? >> >> Lastly, thank you for your patience and for keeping this brain-trust alive. >> I am quite grateful for all the time and energy members pour into this list. >> The archives have been a good source of learning material. >> >> -Randal (at CubeCentral Labs...) >> >> >> >> >> >> >> >> >> >> >> >> >> >> >> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
BK
Bob kb8tq
Thu, Nov 16, 2017 4:15 PM

Hi

Yes, that’s exactly what I’m saying. You just use the software rather than dragging around a
big hunk of coax. It makes it easy to get one pps into the “that’s way more than I need” range.
With the coax approach, is 50NS enough? Might 100NS be needed? Is there a 231NS case?.
I’ve spent a lot of time finding those cases in the middle of long data runs ….

Bob

On Nov 16, 2017, at 10:37 AM, Jerry jsternmd@att.net wrote:

Bob,

Do you mean then you do not need to put a physical long length of cable for the delay, just do it in software or do you do both?

Jerry, NY2KW

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob kb8tq
Sent: Thursday, November 16, 2017 9:58 AM
To: Discussion of precise time and frequency measurement time-nuts@febo.com
Subject: Re: [time-nuts] Interpreting and Understanding Allen Deviation Results

Hi

One “cute” thing to do when looking at GPSDO’s or GPS modules is to use the “cable delay” setting. It will allow you to move the pps of one unit relative to the pps of the other one. You then can be sure of which pps happens first. That makes the A to B measurement much easier to analyze.

Short intervals also can lessen the impact of the time base accuracy in the counter ( you always are measuring a microsecond or so to a nanosecond resolution). Indeed there are other issues (like jitter) that still are an issue.

Bob

On Nov 16, 2017, at 4:10 AM, Azelio Boriani azelio.boriani@gmail.com wrote:

As already stated here, the best measurement mode is the time-interval
mode. The 5335A is a 2ns single-shot resolution counter. Use the PPS
output from the GPSDO, route it to the A (start) input and to a
coaxial cable used as a delay line (10m, 50ns, should be enough). The
other end of the cable into the B input (stop), select the time
interval mode TIME A -> B. Let the internal reference clock the
counter. Set trigger levels and the various parameter to get stable
readings and collect your data.

On Thu, Nov 16, 2017 at 3:59 AM, Mike Garvey r3m1g4@verizon.net wrote:

Could you post some phase plots?  The data you show is not 1/tau and very likely not white phase noise.
Mike

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of
CubeCentral
Sent: Wednesday, November 15, 2017 11:12
To: time-nuts@febo.com
Subject: [time-nuts] Interpreting and Understanding Allen Deviation
Results

Greetings, time-nuts!

After reading [ http://www.leapsecond.com/pages/adev/adev-why.htm ] I felt that I better understood how an Allan Deviation is calculated and endeavored to try an experiment.  It should be noted that I have a hobbyist-level understanding of the concepts described and tools used below.  If my thinking or test methodology is incorrect, please let me know so that I might learn something.

A GPSDO with a 10MHz output was run into the EXT TIME BASE input on the back of an HP5335A.
Then, the TIME BASE OUT on the back was run to the A input on the front of the HP5335A.
My intention was to characterize the performance of the HP5335A counter itself so that I might understand better future plots involving other GPSDO and the counter's internal clock (which was bypassed for this test).

The settings of the HP5335A were as follows:
Gate Mode: Normal
Cycle: Normal

A Input ------------------------------ Trigger Adjust: Full left to
'Preset' detent
Z select  =  in  =  50ohm
x10 ATTN  =  in  =  x10 ATTN  (should have been out/off?)
Slope    =  out  =  up
AC        =  in  =  AC coupled
COMA      =  out  =  Not ComA
AutoTrig  =  out  =  Not Auto Tiggered (should have been in/on?)

(Tangentially, if someone has a good 'primer' or how-to resource
detailing Universal Counter operation, showing when/why/how to set
the knobs in certain situations it would be welcome!)

I then set the Time Lab V1.29 software to repeatedly acquire data for
12 hours, starting the next test as soon as I could.  This means
that, normally, a test was run during the day for 12 hours, and then
overnight for
12 hours.

The results are shown here:  [ https://i.imgur.com/0sMVMfk.png ]  The associated .TIM files are available upon request.

So, now we get to the heart of the matter and the questions this test and results have raised.
I am trying to understand what the data is telling me about the test, and therefore the character of the counter.

  1. Why are the plots a straight line from ~0.25s until ~100s?
  2. Why, after falling at the start, do the plots all seem to go back up from ~100s to ~1000s?
  3. What do the "peaks" mean, after the plot has fallen and begin to rise again?
  4. Why is the period from ~1000s to ~10000s so chaotic?
  5. The pattern "Fall to a minimum point, then rise to a peak, then fall again" seems to be prevalent.  What does that indicate?
  6. Why does that pattern in question (5) seem to repeat sometimes?  What is that showing me?

And finally, some general questions about looking at these plots.
a)  Would a "perfect" plot be a straight line falling from left to right?
(Meaning a hypothetical "ideal" source with perfect timing?)
b)  Is there some example showing plots from two different sources that then describes why one source is better than the other (based upon the ADEV plot)?
c)  I believe that if I understood the math better, these types of plots would be more telling.  Without having to dive back into my college Calculus or Statistics books, is there a good resource for me to be able to understand this better?

Lastly, thank you for your patience and for keeping this brain-trust alive.
I am quite grateful for all the time and energy members pour into this list.
The archives have been a good source of learning material.

   -Randal (at CubeCentral Labs...)

time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Yes, that’s exactly what I’m saying. You just use the software rather than dragging around a big hunk of coax. It makes it easy to get one pps into the “that’s way more than I need” range. With the coax approach, is 50NS enough? Might 100NS be needed? Is there a 231NS case?. I’ve spent a *lot* of time finding those cases in the middle of long data runs …. Bob > On Nov 16, 2017, at 10:37 AM, Jerry <jsternmd@att.net> wrote: > > Bob, > > Do you mean then you do not need to put a physical long length of cable for the delay, just do it in software or do you do both? > > Jerry, NY2KW > > -----Original Message----- > From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob kb8tq > Sent: Thursday, November 16, 2017 9:58 AM > To: Discussion of precise time and frequency measurement <time-nuts@febo.com> > Subject: Re: [time-nuts] Interpreting and Understanding Allen Deviation Results > > Hi > > One “cute” thing to do when looking at GPSDO’s or GPS modules is to use the “cable delay” setting. It will allow you to move the pps of one unit relative to the pps of the other one. You then can be sure of which pps happens first. That makes the A to B measurement much easier to analyze. > > Short intervals also can lessen the impact of the time base accuracy in the counter ( you always are measuring a microsecond or so to a nanosecond resolution). Indeed there are other issues (like jitter) that still are an issue. > > Bob > >> On Nov 16, 2017, at 4:10 AM, Azelio Boriani <azelio.boriani@gmail.com> wrote: >> >> As already stated here, the best measurement mode is the time-interval >> mode. The 5335A is a 2ns single-shot resolution counter. Use the PPS >> output from the GPSDO, route it to the A (start) input and to a >> coaxial cable used as a delay line (10m, 50ns, should be enough). The >> other end of the cable into the B input (stop), select the time >> interval mode TIME A -> B. Let the internal reference clock the >> counter. Set trigger levels and the various parameter to get stable >> readings and collect your data. >> >> On Thu, Nov 16, 2017 at 3:59 AM, Mike Garvey <r3m1g4@verizon.net> wrote: >>> Could you post some phase plots? The data you show is not 1/tau and very likely not white phase noise. >>> Mike >>> >>> -----Original Message----- >>> From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of >>> CubeCentral >>> Sent: Wednesday, November 15, 2017 11:12 >>> To: time-nuts@febo.com >>> Subject: [time-nuts] Interpreting and Understanding Allen Deviation >>> Results >>> >>> Greetings, time-nuts! >>> >>> After reading [ http://www.leapsecond.com/pages/adev/adev-why.htm ] I felt that I better understood how an Allan Deviation is calculated and endeavored to try an experiment. It should be noted that I have a hobbyist-level understanding of the concepts described and tools used below. If my thinking or test methodology is incorrect, please let me know so that I might learn something. >>> >>> A GPSDO with a 10MHz output was run into the EXT TIME BASE input on the back of an HP5335A. >>> Then, the TIME BASE OUT on the back was run to the A input on the front of the HP5335A. >>> My intention was to characterize the performance of the HP5335A counter itself so that I might understand better future plots involving other GPSDO and the counter's internal clock (which was bypassed for this test). >>> >>> The settings of the HP5335A were as follows: >>> Gate Mode: Normal >>> Cycle: Normal >>> >>> A Input ------------------------------ Trigger Adjust: Full left to >>> 'Preset' detent >>> Z select = in = 50ohm >>> x10 ATTN = in = x10 ATTN (should have been out/off?) >>> Slope = out = up >>> AC = in = AC coupled >>> COMA = out = Not ComA >>> AutoTrig = out = Not Auto Tiggered (should have been in/on?) >>> >>> (Tangentially, if someone has a good 'primer' or how-to resource >>> detailing Universal Counter operation, showing when/why/how to set >>> the knobs in certain situations it would be welcome!) >>> >>> I then set the Time Lab V1.29 software to repeatedly acquire data for >>> 12 hours, starting the next test as soon as I could. This means >>> that, normally, a test was run during the day for 12 hours, and then >>> overnight for >>> 12 hours. >>> >>> The results are shown here: [ https://i.imgur.com/0sMVMfk.png ] The associated .TIM files are available upon request. >>> >>> So, now we get to the heart of the matter and the questions this test and results have raised. >>> I am trying to understand what the data is telling me about the test, and therefore the character of the counter. >>> >>> 1) Why are the plots a straight line from ~0.25s until ~100s? >>> 2) Why, after falling at the start, do the plots all seem to go back up from ~100s to ~1000s? >>> 3) What do the "peaks" mean, after the plot has fallen and begin to rise again? >>> 4) Why is the period from ~1000s to ~10000s so chaotic? >>> 5) The pattern "Fall to a minimum point, then rise to a peak, then fall again" seems to be prevalent. What does that indicate? >>> 6) Why does that pattern in question (5) seem to repeat sometimes? What is that showing me? >>> >>> And finally, some general questions about looking at these plots. >>> a) Would a "perfect" plot be a straight line falling from left to right? >>> (Meaning a hypothetical "ideal" source with perfect timing?) >>> b) Is there some example showing plots from two different sources that then describes why one source is better than the other (based upon the ADEV plot)? >>> c) I believe that if I understood the math better, these types of plots would be more telling. Without having to dive back into my college Calculus or Statistics books, is there a good resource for me to be able to understand this better? >>> >>> Lastly, thank you for your patience and for keeping this brain-trust alive. >>> I am quite grateful for all the time and energy members pour into this list. >>> The archives have been a good source of learning material. >>> >>> -Randal (at CubeCentral Labs...) >>> >>> >>> >>> >>> >>> >>> >>> >>> >>> >>> >>> >>> >>> >>> >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
J
Jerry
Thu, Nov 16, 2017 5:05 PM

Bob,

I am also a time newbie... how do you adjust this in software?  For time interval as discussed below, the unaltered GPSDO output goes to A and how do you create the GPSDO delay for B without a physical coax delay?  Any change in GPSDO cable delay setting will affect A and B the same.  Sorry if this is a stupid question

Jerry, NY2KW

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob kb8tq
Sent: Thursday, November 16, 2017 11:15 AM
To: Discussion of precise time and frequency measurement time-nuts@febo.com
Subject: Re: [time-nuts] Interpreting and Understanding Allen Deviation Results

Hi

Yes, that’s exactly what I’m saying. You just use the software rather than dragging around a big hunk of coax. It makes it easy to get one pps into the “that’s way more than I need” range.
With the coax approach, is 50NS enough? Might 100NS be needed? Is there a 231NS case?.
I’ve spent a lot of time finding those cases in the middle of long data runs ….

Bob

On Nov 16, 2017, at 10:37 AM, Jerry jsternmd@att.net wrote:

Bob,

Do you mean then you do not need to put a physical long length of cable for the delay, just do it in software or do you do both?

Jerry, NY2KW

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob
kb8tq
Sent: Thursday, November 16, 2017 9:58 AM
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Subject: Re: [time-nuts] Interpreting and Understanding Allen
Deviation Results

Hi

One “cute” thing to do when looking at GPSDO’s or GPS modules is to use the “cable delay” setting. It will allow you to move the pps of one unit relative to the pps of the other one. You then can be sure of which pps happens first. That makes the A to B measurement much easier to analyze.

Short intervals also can lessen the impact of the time base accuracy in the counter ( you always are measuring a microsecond or so to a nanosecond resolution). Indeed there are other issues (like jitter) that still are an issue.

Bob

On Nov 16, 2017, at 4:10 AM, Azelio Boriani azelio.boriani@gmail.com wrote:

As already stated here, the best measurement mode is the
time-interval mode. The 5335A is a 2ns single-shot resolution
counter. Use the PPS output from the GPSDO, route it to the A (start)
input and to a coaxial cable used as a delay line (10m, 50ns, should
be enough). The other end of the cable into the B input (stop),
select the time interval mode TIME A -> B. Let the internal reference
clock the counter. Set trigger levels and the various parameter to
get stable readings and collect your data.

On Thu, Nov 16, 2017 at 3:59 AM, Mike Garvey r3m1g4@verizon.net wrote:

Could you post some phase plots?  The data you show is not 1/tau and very likely not white phase noise.
Mike

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of
CubeCentral
Sent: Wednesday, November 15, 2017 11:12
To: time-nuts@febo.com
Subject: [time-nuts] Interpreting and Understanding Allen Deviation
Results

Greetings, time-nuts!

After reading [ http://www.leapsecond.com/pages/adev/adev-why.htm ] I felt that I better understood how an Allan Deviation is calculated and endeavored to try an experiment.  It should be noted that I have a hobbyist-level understanding of the concepts described and tools used below.  If my thinking or test methodology is incorrect, please let me know so that I might learn something.

A GPSDO with a 10MHz output was run into the EXT TIME BASE input on the back of an HP5335A.
Then, the TIME BASE OUT on the back was run to the A input on the front of the HP5335A.
My intention was to characterize the performance of the HP5335A counter itself so that I might understand better future plots involving other GPSDO and the counter's internal clock (which was bypassed for this test).

The settings of the HP5335A were as follows:
Gate Mode: Normal
Cycle: Normal

A Input ------------------------------ Trigger Adjust: Full left to
'Preset' detent
Z select  =  in  =  50ohm
x10 ATTN  =  in  =  x10 ATTN  (should have been out/off?)
Slope    =  out  =  up
AC        =  in  =  AC coupled
COMA      =  out  =  Not ComA
AutoTrig  =  out  =  Not Auto Tiggered (should have been in/on?)

(Tangentially, if someone has a good 'primer' or how-to resource
detailing Universal Counter operation, showing when/why/how to set
the knobs in certain situations it would be welcome!)

I then set the Time Lab V1.29 software to repeatedly acquire data
for
12 hours, starting the next test as soon as I could.  This means
that, normally, a test was run during the day for 12 hours, and then
overnight for
12 hours.

The results are shown here:  [ https://i.imgur.com/0sMVMfk.png ]  The associated .TIM files are available upon request.

So, now we get to the heart of the matter and the questions this test and results have raised.
I am trying to understand what the data is telling me about the test, and therefore the character of the counter.

  1. Why are the plots a straight line from ~0.25s until ~100s?
  2. Why, after falling at the start, do the plots all seem to go back up from ~100s to ~1000s?
  3. What do the "peaks" mean, after the plot has fallen and begin to rise again?
  4. Why is the period from ~1000s to ~10000s so chaotic?
  5. The pattern "Fall to a minimum point, then rise to a peak, then fall again" seems to be prevalent.  What does that indicate?
  6. Why does that pattern in question (5) seem to repeat sometimes?  What is that showing me?

And finally, some general questions about looking at these plots.
a)  Would a "perfect" plot be a straight line falling from left to right?
(Meaning a hypothetical "ideal" source with perfect timing?)
b)  Is there some example showing plots from two different sources that then describes why one source is better than the other (based upon the ADEV plot)?
c)  I believe that if I understood the math better, these types of plots would be more telling.  Without having to dive back into my college Calculus or Statistics books, is there a good resource for me to be able to understand this better?

Lastly, thank you for your patience and for keeping this brain-trust alive.
I am quite grateful for all the time and energy members pour into this list.
The archives have been a good source of learning material.

   -Randal (at CubeCentral Labs...)

time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Bob, I am also a time newbie... how do you adjust this in software? For time interval as discussed below, the unaltered GPSDO output goes to A and how do you create the GPSDO delay for B without a physical coax delay? Any change in GPSDO cable delay setting will affect A and B the same. Sorry if this is a stupid question Jerry, NY2KW -----Original Message----- From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob kb8tq Sent: Thursday, November 16, 2017 11:15 AM To: Discussion of precise time and frequency measurement <time-nuts@febo.com> Subject: Re: [time-nuts] Interpreting and Understanding Allen Deviation Results Hi Yes, that’s exactly what I’m saying. You just use the software rather than dragging around a big hunk of coax. It makes it easy to get one pps into the “that’s way more than I need” range. With the coax approach, is 50NS enough? Might 100NS be needed? Is there a 231NS case?. I’ve spent a *lot* of time finding those cases in the middle of long data runs …. Bob > On Nov 16, 2017, at 10:37 AM, Jerry <jsternmd@att.net> wrote: > > Bob, > > Do you mean then you do not need to put a physical long length of cable for the delay, just do it in software or do you do both? > > Jerry, NY2KW > > -----Original Message----- > From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob > kb8tq > Sent: Thursday, November 16, 2017 9:58 AM > To: Discussion of precise time and frequency measurement > <time-nuts@febo.com> > Subject: Re: [time-nuts] Interpreting and Understanding Allen > Deviation Results > > Hi > > One “cute” thing to do when looking at GPSDO’s or GPS modules is to use the “cable delay” setting. It will allow you to move the pps of one unit relative to the pps of the other one. You then can be sure of which pps happens first. That makes the A to B measurement much easier to analyze. > > Short intervals also can lessen the impact of the time base accuracy in the counter ( you always are measuring a microsecond or so to a nanosecond resolution). Indeed there are other issues (like jitter) that still are an issue. > > Bob > >> On Nov 16, 2017, at 4:10 AM, Azelio Boriani <azelio.boriani@gmail.com> wrote: >> >> As already stated here, the best measurement mode is the >> time-interval mode. The 5335A is a 2ns single-shot resolution >> counter. Use the PPS output from the GPSDO, route it to the A (start) >> input and to a coaxial cable used as a delay line (10m, 50ns, should >> be enough). The other end of the cable into the B input (stop), >> select the time interval mode TIME A -> B. Let the internal reference >> clock the counter. Set trigger levels and the various parameter to >> get stable readings and collect your data. >> >> On Thu, Nov 16, 2017 at 3:59 AM, Mike Garvey <r3m1g4@verizon.net> wrote: >>> Could you post some phase plots? The data you show is not 1/tau and very likely not white phase noise. >>> Mike >>> >>> -----Original Message----- >>> From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of >>> CubeCentral >>> Sent: Wednesday, November 15, 2017 11:12 >>> To: time-nuts@febo.com >>> Subject: [time-nuts] Interpreting and Understanding Allen Deviation >>> Results >>> >>> Greetings, time-nuts! >>> >>> After reading [ http://www.leapsecond.com/pages/adev/adev-why.htm ] I felt that I better understood how an Allan Deviation is calculated and endeavored to try an experiment. It should be noted that I have a hobbyist-level understanding of the concepts described and tools used below. If my thinking or test methodology is incorrect, please let me know so that I might learn something. >>> >>> A GPSDO with a 10MHz output was run into the EXT TIME BASE input on the back of an HP5335A. >>> Then, the TIME BASE OUT on the back was run to the A input on the front of the HP5335A. >>> My intention was to characterize the performance of the HP5335A counter itself so that I might understand better future plots involving other GPSDO and the counter's internal clock (which was bypassed for this test). >>> >>> The settings of the HP5335A were as follows: >>> Gate Mode: Normal >>> Cycle: Normal >>> >>> A Input ------------------------------ Trigger Adjust: Full left to >>> 'Preset' detent >>> Z select = in = 50ohm >>> x10 ATTN = in = x10 ATTN (should have been out/off?) >>> Slope = out = up >>> AC = in = AC coupled >>> COMA = out = Not ComA >>> AutoTrig = out = Not Auto Tiggered (should have been in/on?) >>> >>> (Tangentially, if someone has a good 'primer' or how-to resource >>> detailing Universal Counter operation, showing when/why/how to set >>> the knobs in certain situations it would be welcome!) >>> >>> I then set the Time Lab V1.29 software to repeatedly acquire data >>> for >>> 12 hours, starting the next test as soon as I could. This means >>> that, normally, a test was run during the day for 12 hours, and then >>> overnight for >>> 12 hours. >>> >>> The results are shown here: [ https://i.imgur.com/0sMVMfk.png ] The associated .TIM files are available upon request. >>> >>> So, now we get to the heart of the matter and the questions this test and results have raised. >>> I am trying to understand what the data is telling me about the test, and therefore the character of the counter. >>> >>> 1) Why are the plots a straight line from ~0.25s until ~100s? >>> 2) Why, after falling at the start, do the plots all seem to go back up from ~100s to ~1000s? >>> 3) What do the "peaks" mean, after the plot has fallen and begin to rise again? >>> 4) Why is the period from ~1000s to ~10000s so chaotic? >>> 5) The pattern "Fall to a minimum point, then rise to a peak, then fall again" seems to be prevalent. What does that indicate? >>> 6) Why does that pattern in question (5) seem to repeat sometimes? What is that showing me? >>> >>> And finally, some general questions about looking at these plots. >>> a) Would a "perfect" plot be a straight line falling from left to right? >>> (Meaning a hypothetical "ideal" source with perfect timing?) >>> b) Is there some example showing plots from two different sources that then describes why one source is better than the other (based upon the ADEV plot)? >>> c) I believe that if I understood the math better, these types of plots would be more telling. Without having to dive back into my college Calculus or Statistics books, is there a good resource for me to be able to understand this better? >>> >>> Lastly, thank you for your patience and for keeping this brain-trust alive. >>> I am quite grateful for all the time and energy members pour into this list. >>> The archives have been a good source of learning material. >>> >>> -Randal (at CubeCentral Labs...) >>> >>> >>> >>> >>> >>> >>> >>> >>> >>> >>> >>> >>> >>> >>> >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.
BK
Bob kb8tq
Thu, Nov 16, 2017 6:18 PM

Hi

The exact software command used depends on the GPSDO you have. How you send the command to
the GPSDO depends a bit on the driver software you are running. Not all GPSDO’s or modules have
cable delay. All the ones I run do have the feature. Normally it comes set to some random default setting
like 60 ns. I typically take one GPSDO and deliberately make it the odd one out of the group. By making
the offset large (say a microsecond or two) it’s easy to spot the “false ticker” in the group if things get
mixed up.

Assuming you set the odd GPSDO to be early (which could be cable delay + or cable delay -) it would
go to input A on the counter. That would be the start channel. Input B would come from any normal
pps signal and it would be the stop channel. You can of course do this other ways that work just as
well.

Again - this is for looking at a pps relative to a GPSDO using a 5335 or similar counter.

Bob

On Nov 16, 2017, at 12:05 PM, Jerry jsternmd@att.net wrote:

Bob,

I am also a time newbie... how do you adjust this in software?  For time interval as discussed below, the unaltered GPSDO output goes to A and how do you create the GPSDO delay for B without a physical coax delay?  Any change in GPSDO cable delay setting will affect A and B the same.  Sorry if this is a stupid question

Jerry, NY2KW

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob kb8tq
Sent: Thursday, November 16, 2017 11:15 AM
To: Discussion of precise time and frequency measurement time-nuts@febo.com
Subject: Re: [time-nuts] Interpreting and Understanding Allen Deviation Results

Hi

Yes, that’s exactly what I’m saying. You just use the software rather than dragging around a big hunk of coax. It makes it easy to get one pps into the “that’s way more than I need” range.
With the coax approach, is 50NS enough? Might 100NS be needed? Is there a 231NS case?.
I’ve spent a lot of time finding those cases in the middle of long data runs ….

Bob

On Nov 16, 2017, at 10:37 AM, Jerry jsternmd@att.net wrote:

Bob,

Do you mean then you do not need to put a physical long length of cable for the delay, just do it in software or do you do both?

Jerry, NY2KW

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob
kb8tq
Sent: Thursday, November 16, 2017 9:58 AM
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Subject: Re: [time-nuts] Interpreting and Understanding Allen
Deviation Results

Hi

One “cute” thing to do when looking at GPSDO’s or GPS modules is to use the “cable delay” setting. It will allow you to move the pps of one unit relative to the pps of the other one. You then can be sure of which pps happens first. That makes the A to B measurement much easier to analyze.

Short intervals also can lessen the impact of the time base accuracy in the counter ( you always are measuring a microsecond or so to a nanosecond resolution). Indeed there are other issues (like jitter) that still are an issue.

Bob

On Nov 16, 2017, at 4:10 AM, Azelio Boriani azelio.boriani@gmail.com wrote:

As already stated here, the best measurement mode is the
time-interval mode. The 5335A is a 2ns single-shot resolution
counter. Use the PPS output from the GPSDO, route it to the A (start)
input and to a coaxial cable used as a delay line (10m, 50ns, should
be enough). The other end of the cable into the B input (stop),
select the time interval mode TIME A -> B. Let the internal reference
clock the counter. Set trigger levels and the various parameter to
get stable readings and collect your data.

On Thu, Nov 16, 2017 at 3:59 AM, Mike Garvey r3m1g4@verizon.net wrote:

Could you post some phase plots?  The data you show is not 1/tau and very likely not white phase noise.
Mike

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of
CubeCentral
Sent: Wednesday, November 15, 2017 11:12
To: time-nuts@febo.com
Subject: [time-nuts] Interpreting and Understanding Allen Deviation
Results

Greetings, time-nuts!

After reading [ http://www.leapsecond.com/pages/adev/adev-why.htm ] I felt that I better understood how an Allan Deviation is calculated and endeavored to try an experiment.  It should be noted that I have a hobbyist-level understanding of the concepts described and tools used below.  If my thinking or test methodology is incorrect, please let me know so that I might learn something.

A GPSDO with a 10MHz output was run into the EXT TIME BASE input on the back of an HP5335A.
Then, the TIME BASE OUT on the back was run to the A input on the front of the HP5335A.
My intention was to characterize the performance of the HP5335A counter itself so that I might understand better future plots involving other GPSDO and the counter's internal clock (which was bypassed for this test).

The settings of the HP5335A were as follows:
Gate Mode: Normal
Cycle: Normal

A Input ------------------------------ Trigger Adjust: Full left to
'Preset' detent
Z select  =  in  =  50ohm
x10 ATTN  =  in  =  x10 ATTN  (should have been out/off?)
Slope    =  out  =  up
AC        =  in  =  AC coupled
COMA      =  out  =  Not ComA
AutoTrig  =  out  =  Not Auto Tiggered (should have been in/on?)

(Tangentially, if someone has a good 'primer' or how-to resource
detailing Universal Counter operation, showing when/why/how to set
the knobs in certain situations it would be welcome!)

I then set the Time Lab V1.29 software to repeatedly acquire data
for
12 hours, starting the next test as soon as I could.  This means
that, normally, a test was run during the day for 12 hours, and then
overnight for
12 hours.

The results are shown here:  [ https://i.imgur.com/0sMVMfk.png ]  The associated .TIM files are available upon request.

So, now we get to the heart of the matter and the questions this test and results have raised.
I am trying to understand what the data is telling me about the test, and therefore the character of the counter.

  1. Why are the plots a straight line from ~0.25s until ~100s?
  2. Why, after falling at the start, do the plots all seem to go back up from ~100s to ~1000s?
  3. What do the "peaks" mean, after the plot has fallen and begin to rise again?
  4. Why is the period from ~1000s to ~10000s so chaotic?
  5. The pattern "Fall to a minimum point, then rise to a peak, then fall again" seems to be prevalent.  What does that indicate?
  6. Why does that pattern in question (5) seem to repeat sometimes?  What is that showing me?

And finally, some general questions about looking at these plots.
a)  Would a "perfect" plot be a straight line falling from left to right?
(Meaning a hypothetical "ideal" source with perfect timing?)
b)  Is there some example showing plots from two different sources that then describes why one source is better than the other (based upon the ADEV plot)?
c)  I believe that if I understood the math better, these types of plots would be more telling.  Without having to dive back into my college Calculus or Statistics books, is there a good resource for me to be able to understand this better?

Lastly, thank you for your patience and for keeping this brain-trust alive.
I am quite grateful for all the time and energy members pour into this list.
The archives have been a good source of learning material.

  -Randal (at CubeCentral Labs...)

time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi The exact software command used depends on the GPSDO you have. How you send the command to the GPSDO depends a bit on the driver software you are running. Not all GPSDO’s or modules have cable delay. All the ones I run do have the feature. Normally it comes set to some random default setting like 60 ns. I typically take one GPSDO and deliberately make it the odd one out of the group. By making the offset large (say a microsecond or two) it’s easy to spot the “false ticker” in the group if things get mixed up. Assuming you set the odd GPSDO to be early (which could be cable delay + or cable delay -) it would go to input A on the counter. That would be the start channel. Input B would come from any normal pps signal and it would be the stop channel. You can of course do this other ways that work just as well. Again - this is for looking at a pps relative to a GPSDO using a 5335 or similar counter. Bob > On Nov 16, 2017, at 12:05 PM, Jerry <jsternmd@att.net> wrote: > > Bob, > > I am also a time newbie... how do you adjust this in software? For time interval as discussed below, the unaltered GPSDO output goes to A and how do you create the GPSDO delay for B without a physical coax delay? Any change in GPSDO cable delay setting will affect A and B the same. Sorry if this is a stupid question > > Jerry, NY2KW > > -----Original Message----- > From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob kb8tq > Sent: Thursday, November 16, 2017 11:15 AM > To: Discussion of precise time and frequency measurement <time-nuts@febo.com> > Subject: Re: [time-nuts] Interpreting and Understanding Allen Deviation Results > > Hi > > Yes, that’s exactly what I’m saying. You just use the software rather than dragging around a big hunk of coax. It makes it easy to get one pps into the “that’s way more than I need” range. > With the coax approach, is 50NS enough? Might 100NS be needed? Is there a 231NS case?. > I’ve spent a *lot* of time finding those cases in the middle of long data runs …. > > Bob > >> On Nov 16, 2017, at 10:37 AM, Jerry <jsternmd@att.net> wrote: >> >> Bob, >> >> Do you mean then you do not need to put a physical long length of cable for the delay, just do it in software or do you do both? >> >> Jerry, NY2KW >> >> -----Original Message----- >> From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob >> kb8tq >> Sent: Thursday, November 16, 2017 9:58 AM >> To: Discussion of precise time and frequency measurement >> <time-nuts@febo.com> >> Subject: Re: [time-nuts] Interpreting and Understanding Allen >> Deviation Results >> >> Hi >> >> One “cute” thing to do when looking at GPSDO’s or GPS modules is to use the “cable delay” setting. It will allow you to move the pps of one unit relative to the pps of the other one. You then can be sure of which pps happens first. That makes the A to B measurement much easier to analyze. >> >> Short intervals also can lessen the impact of the time base accuracy in the counter ( you always are measuring a microsecond or so to a nanosecond resolution). Indeed there are other issues (like jitter) that still are an issue. >> >> Bob >> >>> On Nov 16, 2017, at 4:10 AM, Azelio Boriani <azelio.boriani@gmail.com> wrote: >>> >>> As already stated here, the best measurement mode is the >>> time-interval mode. The 5335A is a 2ns single-shot resolution >>> counter. Use the PPS output from the GPSDO, route it to the A (start) >>> input and to a coaxial cable used as a delay line (10m, 50ns, should >>> be enough). The other end of the cable into the B input (stop), >>> select the time interval mode TIME A -> B. Let the internal reference >>> clock the counter. Set trigger levels and the various parameter to >>> get stable readings and collect your data. >>> >>> On Thu, Nov 16, 2017 at 3:59 AM, Mike Garvey <r3m1g4@verizon.net> wrote: >>>> Could you post some phase plots? The data you show is not 1/tau and very likely not white phase noise. >>>> Mike >>>> >>>> -----Original Message----- >>>> From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of >>>> CubeCentral >>>> Sent: Wednesday, November 15, 2017 11:12 >>>> To: time-nuts@febo.com >>>> Subject: [time-nuts] Interpreting and Understanding Allen Deviation >>>> Results >>>> >>>> Greetings, time-nuts! >>>> >>>> After reading [ http://www.leapsecond.com/pages/adev/adev-why.htm ] I felt that I better understood how an Allan Deviation is calculated and endeavored to try an experiment. It should be noted that I have a hobbyist-level understanding of the concepts described and tools used below. If my thinking or test methodology is incorrect, please let me know so that I might learn something. >>>> >>>> A GPSDO with a 10MHz output was run into the EXT TIME BASE input on the back of an HP5335A. >>>> Then, the TIME BASE OUT on the back was run to the A input on the front of the HP5335A. >>>> My intention was to characterize the performance of the HP5335A counter itself so that I might understand better future plots involving other GPSDO and the counter's internal clock (which was bypassed for this test). >>>> >>>> The settings of the HP5335A were as follows: >>>> Gate Mode: Normal >>>> Cycle: Normal >>>> >>>> A Input ------------------------------ Trigger Adjust: Full left to >>>> 'Preset' detent >>>> Z select = in = 50ohm >>>> x10 ATTN = in = x10 ATTN (should have been out/off?) >>>> Slope = out = up >>>> AC = in = AC coupled >>>> COMA = out = Not ComA >>>> AutoTrig = out = Not Auto Tiggered (should have been in/on?) >>>> >>>> (Tangentially, if someone has a good 'primer' or how-to resource >>>> detailing Universal Counter operation, showing when/why/how to set >>>> the knobs in certain situations it would be welcome!) >>>> >>>> I then set the Time Lab V1.29 software to repeatedly acquire data >>>> for >>>> 12 hours, starting the next test as soon as I could. This means >>>> that, normally, a test was run during the day for 12 hours, and then >>>> overnight for >>>> 12 hours. >>>> >>>> The results are shown here: [ https://i.imgur.com/0sMVMfk.png ] The associated .TIM files are available upon request. >>>> >>>> So, now we get to the heart of the matter and the questions this test and results have raised. >>>> I am trying to understand what the data is telling me about the test, and therefore the character of the counter. >>>> >>>> 1) Why are the plots a straight line from ~0.25s until ~100s? >>>> 2) Why, after falling at the start, do the plots all seem to go back up from ~100s to ~1000s? >>>> 3) What do the "peaks" mean, after the plot has fallen and begin to rise again? >>>> 4) Why is the period from ~1000s to ~10000s so chaotic? >>>> 5) The pattern "Fall to a minimum point, then rise to a peak, then fall again" seems to be prevalent. What does that indicate? >>>> 6) Why does that pattern in question (5) seem to repeat sometimes? What is that showing me? >>>> >>>> And finally, some general questions about looking at these plots. >>>> a) Would a "perfect" plot be a straight line falling from left to right? >>>> (Meaning a hypothetical "ideal" source with perfect timing?) >>>> b) Is there some example showing plots from two different sources that then describes why one source is better than the other (based upon the ADEV plot)? >>>> c) I believe that if I understood the math better, these types of plots would be more telling. Without having to dive back into my college Calculus or Statistics books, is there a good resource for me to be able to understand this better? >>>> >>>> Lastly, thank you for your patience and for keeping this brain-trust alive. >>>> I am quite grateful for all the time and energy members pour into this list. >>>> The archives have been a good source of learning material. >>>> >>>> -Randal (at CubeCentral Labs...) >>>> >>>> >>>> >>>> >>>> >>>> >>>> >>>> >>>> >>>> >>>> >>>> >>>> >>>> >>>> >>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>> and follow the instructions there. >>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>> and follow the instructions there. >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
AB
Azelio Boriani
Thu, Nov 16, 2017 6:40 PM

If you have only one GPSDO, a coaxial cable is your best friend to
test the performance of the counter in time-interval mode. If you have
two GPSDOs, the Segal's law apply:
https://en.wikipedia.org/wiki/Segal%27s_law, better go directly for
three.

On Thu, Nov 16, 2017 at 7:18 PM, Bob kb8tq kb8tq@n1k.org wrote:

Hi

The exact software command used depends on the GPSDO you have. How you send the command to
the GPSDO depends a bit on the driver software you are running. Not all GPSDO’s or modules have
cable delay. All the ones I run do have the feature. Normally it comes set to some random default setting
like 60 ns. I typically take one GPSDO and deliberately make it the odd one out of the group. By making
the offset large (say a microsecond or two) it’s easy to spot the “false ticker” in the group if things get
mixed up.

Assuming you set the odd GPSDO to be early (which could be cable delay + or cable delay -) it would
go to input A on the counter. That would be the start channel. Input B would come from any normal
pps signal and it would be the stop channel. You can of course do this other ways that work just as
well.

Again - this is for looking at a pps relative to a GPSDO using a 5335 or similar counter.

Bob

On Nov 16, 2017, at 12:05 PM, Jerry jsternmd@att.net wrote:

Bob,

I am also a time newbie... how do you adjust this in software?  For time interval as discussed below, the unaltered GPSDO output goes to A and how do you create the GPSDO delay for B without a physical coax delay?  Any change in GPSDO cable delay setting will affect A and B the same.  Sorry if this is a stupid question

Jerry, NY2KW

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob kb8tq
Sent: Thursday, November 16, 2017 11:15 AM
To: Discussion of precise time and frequency measurement time-nuts@febo.com
Subject: Re: [time-nuts] Interpreting and Understanding Allen Deviation Results

Hi

Yes, that’s exactly what I’m saying. You just use the software rather than dragging around a big hunk of coax. It makes it easy to get one pps into the “that’s way more than I need” range.
With the coax approach, is 50NS enough? Might 100NS be needed? Is there a 231NS case?.
I’ve spent a lot of time finding those cases in the middle of long data runs ….

Bob

On Nov 16, 2017, at 10:37 AM, Jerry jsternmd@att.net wrote:

Bob,

Do you mean then you do not need to put a physical long length of cable for the delay, just do it in software or do you do both?

Jerry, NY2KW

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob
kb8tq
Sent: Thursday, November 16, 2017 9:58 AM
To: Discussion of precise time and frequency measurement
time-nuts@febo.com
Subject: Re: [time-nuts] Interpreting and Understanding Allen
Deviation Results

Hi

One “cute” thing to do when looking at GPSDO’s or GPS modules is to use the “cable delay” setting. It will allow you to move the pps of one unit relative to the pps of the other one. You then can be sure of which pps happens first. That makes the A to B measurement much easier to analyze.

Short intervals also can lessen the impact of the time base accuracy in the counter ( you always are measuring a microsecond or so to a nanosecond resolution). Indeed there are other issues (like jitter) that still are an issue.

Bob

On Nov 16, 2017, at 4:10 AM, Azelio Boriani azelio.boriani@gmail.com wrote:

As already stated here, the best measurement mode is the
time-interval mode. The 5335A is a 2ns single-shot resolution
counter. Use the PPS output from the GPSDO, route it to the A (start)
input and to a coaxial cable used as a delay line (10m, 50ns, should
be enough). The other end of the cable into the B input (stop),
select the time interval mode TIME A -> B. Let the internal reference
clock the counter. Set trigger levels and the various parameter to
get stable readings and collect your data.

On Thu, Nov 16, 2017 at 3:59 AM, Mike Garvey r3m1g4@verizon.net wrote:

Could you post some phase plots?  The data you show is not 1/tau and very likely not white phase noise.
Mike

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of
CubeCentral
Sent: Wednesday, November 15, 2017 11:12
To: time-nuts@febo.com
Subject: [time-nuts] Interpreting and Understanding Allen Deviation
Results

Greetings, time-nuts!

After reading [ http://www.leapsecond.com/pages/adev/adev-why.htm ] I felt that I better understood how an Allan Deviation is calculated and endeavored to try an experiment.  It should be noted that I have a hobbyist-level understanding of the concepts described and tools used below.  If my thinking or test methodology is incorrect, please let me know so that I might learn something.

A GPSDO with a 10MHz output was run into the EXT TIME BASE input on the back of an HP5335A.
Then, the TIME BASE OUT on the back was run to the A input on the front of the HP5335A.
My intention was to characterize the performance of the HP5335A counter itself so that I might understand better future plots involving other GPSDO and the counter's internal clock (which was bypassed for this test).

The settings of the HP5335A were as follows:
Gate Mode: Normal
Cycle: Normal

A Input ------------------------------ Trigger Adjust: Full left to
'Preset' detent
Z select  =  in  =  50ohm
x10 ATTN  =  in  =  x10 ATTN  (should have been out/off?)
Slope    =  out  =  up
AC        =  in  =  AC coupled
COMA      =  out  =  Not ComA
AutoTrig  =  out  =  Not Auto Tiggered (should have been in/on?)

(Tangentially, if someone has a good 'primer' or how-to resource
detailing Universal Counter operation, showing when/why/how to set
the knobs in certain situations it would be welcome!)

I then set the Time Lab V1.29 software to repeatedly acquire data
for
12 hours, starting the next test as soon as I could.  This means
that, normally, a test was run during the day for 12 hours, and then
overnight for
12 hours.

The results are shown here:  [ https://i.imgur.com/0sMVMfk.png ]  The associated .TIM files are available upon request.

So, now we get to the heart of the matter and the questions this test and results have raised.
I am trying to understand what the data is telling me about the test, and therefore the character of the counter.

  1. Why are the plots a straight line from ~0.25s until ~100s?
  2. Why, after falling at the start, do the plots all seem to go back up from ~100s to ~1000s?
  3. What do the "peaks" mean, after the plot has fallen and begin to rise again?
  4. Why is the period from ~1000s to ~10000s so chaotic?
  5. The pattern "Fall to a minimum point, then rise to a peak, then fall again" seems to be prevalent.  What does that indicate?
  6. Why does that pattern in question (5) seem to repeat sometimes?  What is that showing me?

And finally, some general questions about looking at these plots.
a)  Would a "perfect" plot be a straight line falling from left to right?
(Meaning a hypothetical "ideal" source with perfect timing?)
b)  Is there some example showing plots from two different sources that then describes why one source is better than the other (based upon the ADEV plot)?
c)  I believe that if I understood the math better, these types of plots would be more telling.  Without having to dive back into my college Calculus or Statistics books, is there a good resource for me to be able to understand this better?

Lastly, thank you for your patience and for keeping this brain-trust alive.
I am quite grateful for all the time and energy members pour into this list.
The archives have been a good source of learning material.

  -Randal (at CubeCentral Labs...)

time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

If you have only one GPSDO, a coaxial cable is your best friend to test the performance of the counter in time-interval mode. If you have two GPSDOs, the Segal's law apply: <https://en.wikipedia.org/wiki/Segal%27s_law>, better go directly for three. On Thu, Nov 16, 2017 at 7:18 PM, Bob kb8tq <kb8tq@n1k.org> wrote: > Hi > > The exact software command used depends on the GPSDO you have. How you send the command to > the GPSDO depends a bit on the driver software you are running. Not all GPSDO’s or modules have > cable delay. All the ones I run do have the feature. Normally it comes set to some random default setting > like 60 ns. I typically take one GPSDO and deliberately make it the odd one out of the group. By making > the offset large (say a microsecond or two) it’s easy to spot the “false ticker” in the group if things get > mixed up. > > Assuming you set the odd GPSDO to be early (which could be cable delay + or cable delay -) it would > go to input A on the counter. That would be the start channel. Input B would come from any normal > pps signal and it would be the stop channel. You can of course do this other ways that work just as > well. > > Again - this is for looking at a pps relative to a GPSDO using a 5335 or similar counter. > > Bob > >> On Nov 16, 2017, at 12:05 PM, Jerry <jsternmd@att.net> wrote: >> >> Bob, >> >> I am also a time newbie... how do you adjust this in software? For time interval as discussed below, the unaltered GPSDO output goes to A and how do you create the GPSDO delay for B without a physical coax delay? Any change in GPSDO cable delay setting will affect A and B the same. Sorry if this is a stupid question >> >> Jerry, NY2KW >> >> -----Original Message----- >> From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob kb8tq >> Sent: Thursday, November 16, 2017 11:15 AM >> To: Discussion of precise time and frequency measurement <time-nuts@febo.com> >> Subject: Re: [time-nuts] Interpreting and Understanding Allen Deviation Results >> >> Hi >> >> Yes, that’s exactly what I’m saying. You just use the software rather than dragging around a big hunk of coax. It makes it easy to get one pps into the “that’s way more than I need” range. >> With the coax approach, is 50NS enough? Might 100NS be needed? Is there a 231NS case?. >> I’ve spent a *lot* of time finding those cases in the middle of long data runs …. >> >> Bob >> >>> On Nov 16, 2017, at 10:37 AM, Jerry <jsternmd@att.net> wrote: >>> >>> Bob, >>> >>> Do you mean then you do not need to put a physical long length of cable for the delay, just do it in software or do you do both? >>> >>> Jerry, NY2KW >>> >>> -----Original Message----- >>> From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of Bob >>> kb8tq >>> Sent: Thursday, November 16, 2017 9:58 AM >>> To: Discussion of precise time and frequency measurement >>> <time-nuts@febo.com> >>> Subject: Re: [time-nuts] Interpreting and Understanding Allen >>> Deviation Results >>> >>> Hi >>> >>> One “cute” thing to do when looking at GPSDO’s or GPS modules is to use the “cable delay” setting. It will allow you to move the pps of one unit relative to the pps of the other one. You then can be sure of which pps happens first. That makes the A to B measurement much easier to analyze. >>> >>> Short intervals also can lessen the impact of the time base accuracy in the counter ( you always are measuring a microsecond or so to a nanosecond resolution). Indeed there are other issues (like jitter) that still are an issue. >>> >>> Bob >>> >>>> On Nov 16, 2017, at 4:10 AM, Azelio Boriani <azelio.boriani@gmail.com> wrote: >>>> >>>> As already stated here, the best measurement mode is the >>>> time-interval mode. The 5335A is a 2ns single-shot resolution >>>> counter. Use the PPS output from the GPSDO, route it to the A (start) >>>> input and to a coaxial cable used as a delay line (10m, 50ns, should >>>> be enough). The other end of the cable into the B input (stop), >>>> select the time interval mode TIME A -> B. Let the internal reference >>>> clock the counter. Set trigger levels and the various parameter to >>>> get stable readings and collect your data. >>>> >>>> On Thu, Nov 16, 2017 at 3:59 AM, Mike Garvey <r3m1g4@verizon.net> wrote: >>>>> Could you post some phase plots? The data you show is not 1/tau and very likely not white phase noise. >>>>> Mike >>>>> >>>>> -----Original Message----- >>>>> From: time-nuts [mailto:time-nuts-bounces@febo.com] On Behalf Of >>>>> CubeCentral >>>>> Sent: Wednesday, November 15, 2017 11:12 >>>>> To: time-nuts@febo.com >>>>> Subject: [time-nuts] Interpreting and Understanding Allen Deviation >>>>> Results >>>>> >>>>> Greetings, time-nuts! >>>>> >>>>> After reading [ http://www.leapsecond.com/pages/adev/adev-why.htm ] I felt that I better understood how an Allan Deviation is calculated and endeavored to try an experiment. It should be noted that I have a hobbyist-level understanding of the concepts described and tools used below. If my thinking or test methodology is incorrect, please let me know so that I might learn something. >>>>> >>>>> A GPSDO with a 10MHz output was run into the EXT TIME BASE input on the back of an HP5335A. >>>>> Then, the TIME BASE OUT on the back was run to the A input on the front of the HP5335A. >>>>> My intention was to characterize the performance of the HP5335A counter itself so that I might understand better future plots involving other GPSDO and the counter's internal clock (which was bypassed for this test). >>>>> >>>>> The settings of the HP5335A were as follows: >>>>> Gate Mode: Normal >>>>> Cycle: Normal >>>>> >>>>> A Input ------------------------------ Trigger Adjust: Full left to >>>>> 'Preset' detent >>>>> Z select = in = 50ohm >>>>> x10 ATTN = in = x10 ATTN (should have been out/off?) >>>>> Slope = out = up >>>>> AC = in = AC coupled >>>>> COMA = out = Not ComA >>>>> AutoTrig = out = Not Auto Tiggered (should have been in/on?) >>>>> >>>>> (Tangentially, if someone has a good 'primer' or how-to resource >>>>> detailing Universal Counter operation, showing when/why/how to set >>>>> the knobs in certain situations it would be welcome!) >>>>> >>>>> I then set the Time Lab V1.29 software to repeatedly acquire data >>>>> for >>>>> 12 hours, starting the next test as soon as I could. This means >>>>> that, normally, a test was run during the day for 12 hours, and then >>>>> overnight for >>>>> 12 hours. >>>>> >>>>> The results are shown here: [ https://i.imgur.com/0sMVMfk.png ] The associated .TIM files are available upon request. >>>>> >>>>> So, now we get to the heart of the matter and the questions this test and results have raised. >>>>> I am trying to understand what the data is telling me about the test, and therefore the character of the counter. >>>>> >>>>> 1) Why are the plots a straight line from ~0.25s until ~100s? >>>>> 2) Why, after falling at the start, do the plots all seem to go back up from ~100s to ~1000s? >>>>> 3) What do the "peaks" mean, after the plot has fallen and begin to rise again? >>>>> 4) Why is the period from ~1000s to ~10000s so chaotic? >>>>> 5) The pattern "Fall to a minimum point, then rise to a peak, then fall again" seems to be prevalent. What does that indicate? >>>>> 6) Why does that pattern in question (5) seem to repeat sometimes? What is that showing me? >>>>> >>>>> And finally, some general questions about looking at these plots. >>>>> a) Would a "perfect" plot be a straight line falling from left to right? >>>>> (Meaning a hypothetical "ideal" source with perfect timing?) >>>>> b) Is there some example showing plots from two different sources that then describes why one source is better than the other (based upon the ADEV plot)? >>>>> c) I believe that if I understood the math better, these types of plots would be more telling. Without having to dive back into my college Calculus or Statistics books, is there a good resource for me to be able to understand this better? >>>>> >>>>> Lastly, thank you for your patience and for keeping this brain-trust alive. >>>>> I am quite grateful for all the time and energy members pour into this list. >>>>> The archives have been a good source of learning material. >>>>> >>>>> -Randal (at CubeCentral Labs...) >>>>> >>>>> >>>>> >>>>> >>>>> >>>>> >>>>> >>>>> >>>>> >>>>> >>>>> >>>>> >>>>> >>>>> >>>>> >>>>> >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>>> and follow the instructions there. >>>>> >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>>> and follow the instructions there. >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>>> and follow the instructions there. >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to >>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.