RE
Randy Evans
Sun, Dec 31, 2017 5:52 AM
Todd,
The minimal quick test would be greatly appreciated if you have the time.
I know this is above and beyond what I would expect but it would be really
useful to see if there is any hope for the GPIB on this unit. I will do
the service note 13 procedure in the morning.
Thanks again,
Randy
On Sat, Dec 30, 2017 at 8:58 PM, Todd Micallef tmicallef@gmail.com wrote:
Randy,
Short of swapping out the A5 digital board I have no other ideas. If you
don't have a NI or Agilent adapter, I am sure I can find a spare for you to
test.
There is a Service Note 3458A-13A that deals with GPIB communication
failures. There is a test within the document to see if a shorting jumper
needs to be installed.
I can try reducing the code to something minimal for a quick test if you
want. It will take some time in the morning.
Todd
On Sat, Dec 30, 2017 at 11:18 PM, Randy Evans randyevans2688@gmail.com
wrote:
Todd,
Same as before. Still not working. All it does is print out "HP3458A"
the output console and then just sits there.
Randy
On Sat, Dec 30, 2017 at 7:43 PM, Todd Micallef tmicallef@gmail.com
wrote:
Randy,
I just remembered that I changed the duration from 30 minutes to 2
to speed up the test. I forgot to change it back.
Todd
On Sat, Dec 30, 2017 at 10:27 PM, Todd Micallef tmicallef@gmail.com
wrote:
Randy,
I just sent it. I haven't modified Martin's code yet. That can be
once you confirm it works for you.
Todd
On Sat, Dec 30, 2017 at 10:12 PM, randyevans2688 <
Todd,
If you don't mind, just send the code directly to me.
Randy
Sent from my Verizon, Samsung Galaxy smartphone
-------- Original message --------From: Todd Micallef <
tmicallef@gmail.com> Date: 12/30/17 6:40 PM (GMT-08:00) To:
of precise voltage measurement volt-nuts@febo.com Subject: Re:
[volt-nuts] GPIB Issue with 3458A
Randy,
I have not written anything else for the 3458A using EzGPIB. If you
a
National Instruments or Agilent adapters, I have a handful of Python
I
use to measure some of my standards.
If you are reading the temperature from the HP meter, I would think
the problem lies in how the triggering is configured. The triggering
been changed in this script so it may work for you. There was an
additional
command added for disabling the buffer.
The code can be posted here, if possible or I can email it to you
testing.
Todd
On Sat, Dec 30, 2017 at 9:26 PM, randyevans2688 <
Todd, thanks for the help. Where will you post the new code?
BTW, do you have any very simple code that is likely to work. I
interested in seeing if anything other than KE5FX's memory dump
will work on my HP3458A.
Thanks again,
Randy
Sent from my Verizon, Samsung Galaxy smartphone
-------- Original message --------From: Todd Micallef <
Date: 12/30/17 6:13 PM (GMT-08:00) To: Discussion of precise
3458A
Randy,
I found the Prologix adapters and have modified the code in the
script. I am doing a test run now and will post the script when I
certain I didn't make it worse.
Todd
On Sat, Dec 30, 2017 at 7:15 PM, Randy Evans <
Todd,
I have swapped cables, moved the Prologix adapter to both meters
works, HP doesn't), tried Martins code as well as Xdevs's. I
tried another PC yet, maybe the next step.
Here is the code that Martin modified for two meters, as well as
modified Xdev's code per your directions (V1.02).
Thanks,
Randy
On Sat, Dec 30, 2017 at 3:55 PM, Todd Micallef <
Randy,
I will continue to look for my adapter. I haven't used it for
while.
I have seen similar issues and swapped cables, moved the
other meter, etc... One of my cables was bad from one side of
to the other (on the same end of the cable) and caused me lots
time. Weird reading errors were happening.
Would you email me Martin's code so I can test it here and I
anything can be done to replicate your issue.
Todd
On Sat, Dec 30, 2017 at 6:42 PM, Randy Evans <
Todd,
I tried adding the line but it made no difference.
Randy
On Sat, Dec 30, 2017 at 12:19 PM, Randy Evans <
Todd,
Does it matter where I put the line in the Init_Device
procedure Init_Device;
begin;
EZGPIB_BusWriteData(gpib_address,GET_ID);
if EZGPIB_BusWaitForData(gpib_
address,meter_id,Timeout)
EZGPIB_ScreenWriteLn(meter_id);
EZGPIB_BusWriteData(gpib_address,TRIGGER_SOURCE);
until meter can be setup
EZGPIB_BusWriteData(gpib_address,'INBUF ON');
EZGPIB_BusWriteData(gpib_address,AUTO_ZERO);
EZGPIB_BusWriteData(gpib_address,OUTPUT_FORMAT);
end;
Randy
On Sat, Dec 30, 2017 at 11:20 AM, Todd Micallef <
Randy,
I am the person that wrote that EzGPIB script. It was my
moved on to Python and I have only written a couple for
meters I have on hand. Version 1.01 is the most current
the 3458A.
I see that I should have added at least one more command
initialization. The additional init command is used in
xDevs.
Add the following line in the Init_Device procedure
EZGPIB_BusWriteData(gpib_address,'END ALWAYS');
This tells the meter to send an EOI on every read.
I am looking for one of my Prologix adapters or I would
find it, I will also test it. I want to make sure there
issues.
Also, do both meters run the same firmware versions? I
meter is using a much older version.
Todd
On Sat, Dec 30, 2017 at 10:42 AM, Randy Evans <
Todd,
I have version 1.01. Yes, please e-mail me your
will help.
Thanks,
Randy
On Sat, Dec 30, 2017 at 4:24 AM, Todd Micallef <
Randy,
Is there a version number at the top of the file? I
and 1.01). I can't remember all that changed between
been a while but I did find an additional timeout
wait time for reading data to 1 minute. I am
mine and added it.
I can email it to you if you want to try it. If it
have Illya update it on xDevs.
Todd
On Sat, Dec 30, 2017 at 12:39 AM, Randy Evans <
I am still unable to get my HP-3458A to work with
"HP3458A_Measure_DMM_Noise.488" (
It
works fine with my Agilent 3458A and with a
Agilent 3458A using the EZGPIB program. The SW
my friend's DMMs, the HP units are (8.2) and with
(9.2). However, both of my units work fine with
HP3458.EXE program to download the NVRAM CAL data.
indicate that my HP-3458 GPIB works at least
Does anyone have a EZGPIB program that I could try
different from the Xdev program)? It would be nice
something to work other than the HP3458.EXE
some ideas on what to look for.
Thanks,
Randy Evans
On Tue, Dec 26, 2017 at 7:24 PM, Randy Evans <
I am hoping some of you experts might be able to
problem i am having trying to make automated
an Agilent and an HP 3458A. I have a program
Reynolds, which is a variant on the Xdevs noise
Martin's program continuously reads the voltage
from the two meters and then stores then in a
the temperature measurements from both meters.
his Agilent and HP 3458As.
On my two 3458As, it takes the temperature
takes a voltage measurement from the Agilent
make any voltage measurements although I can see
states and, of course, it does sent the
if I turn the HP unit off and then on again, the
voltage and temperature measurements and send
Any ideas on what might be causing the strange
mailman/listinfo/volt-nuts
and follow the instructions there.
Todd,
The minimal quick test would be greatly appreciated if you have the time.
I know this is above and beyond what I would expect but it would be really
useful to see if there is any hope for the GPIB on this unit. I will do
the service note 13 procedure in the morning.
Thanks again,
Randy
On Sat, Dec 30, 2017 at 8:58 PM, Todd Micallef <tmicallef@gmail.com> wrote:
> Randy,
>
> Short of swapping out the A5 digital board I have no other ideas. If you
> don't have a NI or Agilent adapter, I am sure I can find a spare for you to
> test.
>
> There is a Service Note 3458A-13A that deals with GPIB communication
> failures. There is a test within the document to see if a shorting jumper
> needs to be installed.
> I can try reducing the code to something minimal for a quick test if you
> want. It will take some time in the morning.
>
> Todd
>
> On Sat, Dec 30, 2017 at 11:18 PM, Randy Evans <randyevans2688@gmail.com>
> wrote:
>
> > Todd,
> >
> > Same as before. Still not working. All it does is print out "HP3458A"
> on
> > the output console and then just sits there.
> >
> > Randy
> >
> > On Sat, Dec 30, 2017 at 7:43 PM, Todd Micallef <tmicallef@gmail.com>
> > wrote:
> >
> > > Randy,
> > >
> > > I just remembered that I changed the duration from 30 minutes to 2
> > minutes
> > > to speed up the test. I forgot to change it back.
> > >
> > > Todd
> > >
> > > On Sat, Dec 30, 2017 at 10:27 PM, Todd Micallef <tmicallef@gmail.com>
> > > wrote:
> > >
> > > > Randy,
> > > >
> > > > I just sent it. I haven't modified Martin's code yet. That can be
> done
> > > > once you confirm it works for you.
> > > >
> > > > Todd
> > > >
> > > > On Sat, Dec 30, 2017 at 10:12 PM, randyevans2688 <
> > > randyevans2688@gmail.com
> > > > > wrote:
> > > >
> > > >> Todd,
> > > >> If you don't mind, just send the code directly to me.
> > > >> Randy
> > > >>
> > > >> Sent from my Verizon, Samsung Galaxy smartphone
> > > >> -------- Original message --------From: Todd Micallef <
> > > >> tmicallef@gmail.com> Date: 12/30/17 6:40 PM (GMT-08:00) To:
> > > Discussion
> > > >> of precise voltage measurement <volt-nuts@febo.com> Subject: Re:
> > > >> [volt-nuts] GPIB Issue with 3458A
> > > >> Randy,
> > > >>
> > > >> I have not written anything else for the 3458A using EzGPIB. If you
> > > have
> > > >> a
> > > >> National Instruments or Agilent adapters, I have a handful of Python
> > > code
> > > >> I
> > > >> use to measure some of my standards.
> > > >>
> > > >> If you are reading the temperature from the HP meter, I would think
> > that
> > > >> the problem lies in how the triggering is configured. The triggering
> > has
> > > >> been changed in this script so it may work for you. There was an
> > > >> additional
> > > >> command added for disabling the buffer.
> > > >> The code can be posted here, if possible or I can email it to you
> for
> > > >> testing.
> > > >>
> > > >> Todd
> > > >>
> > > >> On Sat, Dec 30, 2017 at 9:26 PM, randyevans2688 <
> > > randyevans2688@gmail.com
> > > >> >
> > > >> wrote:
> > > >>
> > > >> > Todd, thanks for the help. Where will you post the new code?
> > > >> > BTW, do you have any very simple code that is likely to work. I
> am
> > > >> > interested in seeing if anything other than KE5FX's memory dump
> > > program
> > > >> > will work on my HP3458A.
> > > >> > Thanks again,
> > > >> > Randy
> > > >> >
> > > >> > Sent from my Verizon, Samsung Galaxy smartphone
> > > >> > -------- Original message --------From: Todd Micallef <
> > > >> tmicallef@gmail.com>
> > > >> > Date: 12/30/17 6:13 PM (GMT-08:00) To: Discussion of precise
> > voltage
> > > >> > measurement <volt-nuts@febo.com> Subject: Re: [volt-nuts] GPIB
> > Issue
> > > >> with
> > > >> > 3458A
> > > >> > Randy,
> > > >> >
> > > >> > I found the Prologix adapters and have modified the code in the
> > > original
> > > >> > script. I am doing a test run now and will post the script when I
> am
> > > >> > certain I didn't make it worse.
> > > >> >
> > > >> > Todd
> > > >> >
> > > >> > On Sat, Dec 30, 2017 at 7:15 PM, Randy Evans <
> > > randyevans2688@gmail.com>
> > > >> > wrote:
> > > >> >
> > > >> > > Todd,
> > > >> > >
> > > >> > > I have swapped cables, moved the Prologix adapter to both meters
> > > >> (Agilent
> > > >> > > works, HP doesn't), tried Martins code as well as Xdevs's. I
> have
> > > not
> > > >> > > tried another PC yet, maybe the next step.
> > > >> > >
> > > >> > > Here is the code that Martin modified for two meters, as well as
> > the
> > > >> > > modified Xdev's code per your directions (V1.02).
> > > >> > >
> > > >> > > Thanks,
> > > >> > >
> > > >> > > Randy
> > > >> > >
> > > >> > > On Sat, Dec 30, 2017 at 3:55 PM, Todd Micallef <
> > tmicallef@gmail.com
> > > >
> > > >> > > wrote:
> > > >> > >
> > > >> > > > Randy,
> > > >> > > >
> > > >> > > > I will continue to look for my adapter. I haven't used it for
> > > quite
> > > >> a
> > > >> > > > while.
> > > >> > > >
> > > >> > > > I have seen similar issues and swapped cables, moved the
> adapter
> > > to
> > > >> the
> > > >> > > > other meter, etc... One of my cables was bad from one side of
> > the
> > > >> > > connector
> > > >> > > > to the other (on the same end of the cable) and caused me lots
> > of
> > > >> > wasted
> > > >> > > > time. Weird reading errors were happening.
> > > >> > > >
> > > >> > > > Would you email me Martin's code so I can test it here and I
> > will
> > > >> see
> > > >> > if
> > > >> > > > anything can be done to replicate your issue.
> > > >> > > >
> > > >> > > > Todd
> > > >> > > >
> > > >> > > >
> > > >> > > > On Sat, Dec 30, 2017 at 6:42 PM, Randy Evans <
> > > >> randyevans2688@gmail.com
> > > >> > >
> > > >> > > > wrote:
> > > >> > > >
> > > >> > > > > Todd,
> > > >> > > > >
> > > >> > > > > I tried adding the line but it made no difference.
> > > >> > > > >
> > > >> > > > > Randy
> > > >> > > > >
> > > >> > > > > On Sat, Dec 30, 2017 at 12:19 PM, Randy Evans <
> > > >> > > randyevans2688@gmail.com>
> > > >> > > > > wrote:
> > > >> > > > >
> > > >> > > > > > Todd,
> > > >> > > > > >
> > > >> > > > > > Does it matter where I put the line in the Init_Device
> > > >> procedure?
> > > >> > > > > >
> > > >> > > > > >
> > > >> > > > > > procedure Init_Device;
> > > >> > > > > > begin;
> > > >> > > > > > EZGPIB_BusWriteData(gpib_address,GET_ID);
> //
> > > Get
> > > >> ID
> > > >> > of
> > > >> > > > > meter
> > > >> > > > > > if EZGPIB_BusWaitForData(gpib_
> address,meter_id,Timeout)
> > > then
> > > >> > > > > > EZGPIB_ScreenWriteLn(meter_id);
> > > >> > > > > >
> > > >> > > > > > EZGPIB_BusWriteData(gpib_address,TRIGGER_SOURCE);
> //
> > > Hold
> > > >> > > > trigger
> > > >> > > > > > until meter can be setup
> > > >> > > > > > EZGPIB_BusWriteData(gpib_address,'INBUF ON');
> > > >> > > > > > EZGPIB_BusWriteData(gpib_address,AUTO_ZERO);
> > > >> > > > > > EZGPIB_BusWriteData(gpib_address,OUTPUT_FORMAT);
> > > >> > > > > > end;
> > > >> > > > > >
> > > >> > > > > >
> > > >> > > > > > Randy
> > > >> > > > > >
> > > >> > > > > > On Sat, Dec 30, 2017 at 11:20 AM, Todd Micallef <
> > > >> > tmicallef@gmail.com
> > > >> > > >
> > > >> > > > > > wrote:
> > > >> > > > > >
> > > >> > > > > >> Randy,
> > > >> > > > > >>
> > > >> > > > > >> I am the person that wrote that EzGPIB script. It was my
> > > first
> > > >> > > before
> > > >> > > > I
> > > >> > > > > >> moved on to Python and I have only written a couple for
> the
> > > >> > > different
> > > >> > > > > >> meters I have on hand. Version 1.01 is the most current
> > > >> version I
> > > >> > > have
> > > >> > > > > for
> > > >> > > > > >> the 3458A.
> > > >> > > > > >> I see that I should have added at least one more command
> to
> > > the
> > > >> > > meter
> > > >> > > > > >> initialization. The additional init command is used in
> the
> > > >> > scripts
> > > >> > > on
> > > >> > > > > >> xDevs.
> > > >> > > > > >>
> > > >> > > > > >> Add the following line in the Init_Device procedure
> > > >> > > > > >> EZGPIB_BusWriteData(gpib_address,'END ALWAYS');
> > > >> > > > > >>
> > > >> > > > > >> This tells the meter to send an EOI on every read.
> > > >> > > > > >>
> > > >> > > > > >> I am looking for one of my Prologix adapters or I would
> > test
> > > >> this.
> > > >> > > > When
> > > >> > > > > I
> > > >> > > > > >> find it, I will also test it. I want to make sure there
> are
> > > no
> > > >> > > further
> > > >> > > > > >> issues.
> > > >> > > > > >> Also, do both meters run the same firmware versions? I
> > wonder
> > > >> if
> > > >> > the
> > > >> > > > HP
> > > >> > > > > >> meter is using a much older version.
> > > >> > > > > >>
> > > >> > > > > >> Todd
> > > >> > > > > >>
> > > >> > > > > >> On Sat, Dec 30, 2017 at 10:42 AM, Randy Evans <
> > > >> > > > randyevans2688@gmail.com
> > > >> > > > > >
> > > >> > > > > >> wrote:
> > > >> > > > > >>
> > > >> > > > > >> > Todd,
> > > >> > > > > >> >
> > > >> > > > > >> > I have version 1.01. Yes, please e-mail me your
> modified
> > > >> > version.
> > > >> > > > > >> Maybe it
> > > >> > > > > >> > will help.
> > > >> > > > > >> >
> > > >> > > > > >> > Thanks,
> > > >> > > > > >> >
> > > >> > > > > >> > Randy
> > > >> > > > > >> >
> > > >> > > > > >> > On Sat, Dec 30, 2017 at 4:24 AM, Todd Micallef <
> > > >> > > tmicallef@gmail.com
> > > >> > > > >
> > > >> > > > > >> > wrote:
> > > >> > > > > >> >
> > > >> > > > > >> > > Randy,
> > > >> > > > > >> > >
> > > >> > > > > >> > > Is there a version number at the top of the file? I
> had
> > > two
> > > >> > > > versions
> > > >> > > > > >> > (1.00
> > > >> > > > > >> > > and 1.01). I can't remember all that changed between
> > the
> > > >> two
> > > >> > > since
> > > >> > > > > it
> > > >> > > > > >> has
> > > >> > > > > >> > > been a while but I did find an additional timeout
> code
> > > that
> > > >> > > > limited
> > > >> > > > > >> the
> > > >> > > > > >> > > wait time for reading data to 1 minute. I am
> guessing I
> > > >> had a
> > > >> > > > > problem
> > > >> > > > > >> > with
> > > >> > > > > >> > > mine and added it.
> > > >> > > > > >> > >
> > > >> > > > > >> > > I can email it to you if you want to try it. If it
> > works
> > > >> for
> > > >> > > you,
> > > >> > > > I
> > > >> > > > > >> will
> > > >> > > > > >> > > have Illya update it on xDevs.
> > > >> > > > > >> > >
> > > >> > > > > >> > > Todd
> > > >> > > > > >> > >
> > > >> > > > > >> > > On Sat, Dec 30, 2017 at 12:39 AM, Randy Evans <
> > > >> > > > > >> randyevans2688@gmail.com>
> > > >> > > > > >> > > wrote:
> > > >> > > > > >> > >
> > > >> > > > > >> > > > I am still unable to get my HP-3458A to work with
> the
> > > >> xdevs
> > > >> > > > > program
> > > >> > > > > >> > > > "HP3458A_Measure_DMM_Noise.488" (
> > > >> > > https://xdevs.com/article/dmm
> > > >> > > > > >> _noise/
> > > >> > > > > >> > ).
> > > >> > > > > >> > > > It
> > > >> > > > > >> > > > works fine with my Agilent 3458A and with a
> friend's
> > > >> > HP-3458A
> > > >> > > > and
> > > >> > > > > >> his
> > > >> > > > > >> > > > Agilent 3458A using the EZGPIB program. The SW
> > > versions
> > > >> are
> > > >> > > > > >> identical
> > > >> > > > > >> > > with
> > > >> > > > > >> > > > my friend's DMMs, the HP units are (8.2) and with
> the
> > > >> > Agilent
> > > >> > > > > units
> > > >> > > > > >> are
> > > >> > > > > >> > > > (9.2). However, both of my units work fine with
> the
> > > >> KE5FX
> > > >> > > GPIB
> > > >> > > > > >> Toolkit
> > > >> > > > > >> > > > HP3458.EXE program to download the NVRAM CAL data.
> > > This
> > > >> > would
> > > >> > > > > seem
> > > >> > > > > >> to
> > > >> > > > > >> > > > indicate that my HP-3458 GPIB works at least
> > partially.
> > > >> > > > > >> > > >
> > > >> > > > > >> > > > Does anyone have a EZGPIB program that I could try
> to
> > > run
> > > >> > > (that
> > > >> > > > is
> > > >> > > > > >> > > > different from the Xdev program)? It would be nice
> > to
> > > >> see
> > > >> > if
> > > >> > > I
> > > >> > > > > >> could
> > > >> > > > > >> > get
> > > >> > > > > >> > > > something to work other than the HP3458.EXE
> program.
> > > Or
> > > >> if
> > > >> > > any
> > > >> > > > > one
> > > >> > > > > >> has
> > > >> > > > > >> > > > some ideas on what to look for.
> > > >> > > > > >> > > >
> > > >> > > > > >> > > > Thanks,
> > > >> > > > > >> > > >
> > > >> > > > > >> > > > Randy Evans
> > > >> > > > > >> > > >
> > > >> > > > > >> > > > On Tue, Dec 26, 2017 at 7:24 PM, Randy Evans <
> > > >> > > > > >> randyevans2688@gmail.com
> > > >> > > > > >> > >
> > > >> > > > > >> > > > wrote:
> > > >> > > > > >> > > >
> > > >> > > > > >> > > > > I am hoping some of you experts might be able to
> > > assist
> > > >> > > with a
> > > >> > > > > >> > strange
> > > >> > > > > >> > > > > problem i am having trying to make automated
> > > >> measurements
> > > >> > > > using
> > > >> > > > > >> > EZGPIB
> > > >> > > > > >> > > > with
> > > >> > > > > >> > > > > an Agilent and an HP 3458A. I have a program
> given
> > > to
> > > >> me
> > > >> > by
> > > >> > > > > >> Martin
> > > >> > > > > >> > > > > Reynolds, which is a variant on the Xdevs noise
> > > >> > measurement
> > > >> > > > > >> program.
> > > >> > > > > >> > > > > Martin's program continuously reads the voltage
> > > >> > measurements
> > > >> > > > > >> > > sequentially
> > > >> > > > > >> > > > > from the two meters and then stores then in a
> > > >> measurements
> > > >> > > > file,
> > > >> > > > > >> > along
> > > >> > > > > >> > > > with
> > > >> > > > > >> > > > > the temperature measurements from both meters.
> The
> > > >> > program
> > > >> > > > > works
> > > >> > > > > >> > fine
> > > >> > > > > >> > > on
> > > >> > > > > >> > > > > his Agilent and HP 3458As.
> > > >> > > > > >> > > > >
> > > >> > > > > >> > > > > On my two 3458As, it takes the temperature
> > > measurements
> > > >> > just
> > > >> > > > > fine
> > > >> > > > > >> but
> > > >> > > > > >> > > > only
> > > >> > > > > >> > > > > takes a voltage measurement from the Agilent
> unit.
> > > >> The HP
> > > >> > > > unit
> > > >> > > > > >> will
> > > >> > > > > >> > > not
> > > >> > > > > >> > > > > make any voltage measurements although I can see
> > both
> > > >> the
> > > >> > > > 3458As
> > > >> > > > > >> in
> > > >> > > > > >> > > talk
> > > >> > > > > >> > > > > states and, of course, it does sent the
> temperature
> > > >> > > > > measurements.
> > > >> > > > > >> > > > However,
> > > >> > > > > >> > > > > if I turn the HP unit off and then on again, the
> HP
> > > >> will
> > > >> > > start
> > > >> > > > > to
> > > >> > > > > >> > take
> > > >> > > > > >> > > > > voltage and temperature measurements and send
> them
> > to
> > > >> the
> > > >> > > PC.
> > > >> > > > > >> > > > >
> > > >> > > > > >> > > > > Any ideas on what might be causing the strange
> > > >> behavior?
> > > >> > > > > >> > > > >
> > > >> > > > > >> > > > > Thanks,
> > > >> > > > > >> > > > >
> > > >> > > > > >> > > > > Randy Evans
> > > >> > > > > >> > > > >
> > > >> > > > > >> > > > _______________________________________________
> > > >> > > > > >> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > >> > > > > >> > > > To unsubscribe, go to
> https://www.febo.com/cgi-bin/
> > > >> > > > > >> > > > mailman/listinfo/volt-nuts
> > > >> > > > > >> > > > and follow the instructions there.
> > > >> > > > > >> > > >
> > > >> > > > > >> > > _______________________________________________
> > > >> > > > > >> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > >> > > > > >> > > To unsubscribe, go to https://www.febo.com/cgi-bin/
> > > >> > > > > >> > > mailman/listinfo/volt-nuts
> > > >> > > > > >> > > and follow the instructions there.
> > > >> > > > > >> > >
> > > >> > > > > >> > _______________________________________________
> > > >> > > > > >> > volt-nuts mailing list -- volt-nuts@febo.com
> > > >> > > > > >> > To unsubscribe, go to https://www.febo.com/cgi-bin/
> > > >> > > > > >> > mailman/listinfo/volt-nuts
> > > >> > > > > >> > and follow the instructions there.
> > > >> > > > > >> >
> > > >> > > > > >> _______________________________________________
> > > >> > > > > >> volt-nuts mailing list -- volt-nuts@febo.com
> > > >> > > > > >> To unsubscribe, go to https://www.febo.com/cgi-bin/m
> > > >> > > > > >> ailman/listinfo/volt-nuts
> > > >> > > > > >> and follow the instructions there.
> > > >> > > > > >>
> > > >> > > > > >
> > > >> > > > > >
> > > >> > > > > _______________________________________________
> > > >> > > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > >> > > > > To unsubscribe, go to https://www.febo.com/cgi-bin/
> > > >> > > > > mailman/listinfo/volt-nuts
> > > >> > > > > and follow the instructions there.
> > > >> > > > >
> > > >> > > > _______________________________________________
> > > >> > > > volt-nuts mailing list -- volt-nuts@febo.com
> > > >> > > > To unsubscribe, go to https://www.febo.com/cgi-bin/
> > > >> > > > mailman/listinfo/volt-nuts
> > > >> > > > and follow the instructions there.
> > > >> > > >
> > > >> > >
> > > >> > > _______________________________________________
> > > >> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > >> > > To unsubscribe, go to https://www.febo.com/cgi-bin/
> > > >> > > mailman/listinfo/volt-nuts
> > > >> > > and follow the instructions there.
> > > >> > >
> > > >> > _______________________________________________
> > > >> > volt-nuts mailing list -- volt-nuts@febo.com
> > > >> > To unsubscribe, go to https://www.febo.com/cgi-bin/
> > > >> > mailman/listinfo/volt-nuts
> > > >> > and follow the instructions there.
> > > >> > _______________________________________________
> > > >> > volt-nuts mailing list -- volt-nuts@febo.com
> > > >> > To unsubscribe, go to https://www.febo.com/cgi-bin/
> > > >> > mailman/listinfo/volt-nuts
> > > >> > and follow the instructions there.
> > > >> >
> > > >> _______________________________________________
> > > >> volt-nuts mailing list -- volt-nuts@febo.com
> > > >> To unsubscribe, go to https://www.febo.com/cgi-bin/m
> > > >> ailman/listinfo/volt-nuts
> > > >> and follow the instructions there.
> > > >> _______________________________________________
> > > >> volt-nuts mailing list -- volt-nuts@febo.com
> > > >> To unsubscribe, go to https://www.febo.com/cgi-bin/m
> > > >> ailman/listinfo/volt-nuts
> > > >> and follow the instructions there.
> > > >>
> > > >
> > > >
> > > _______________________________________________
> > > volt-nuts mailing list -- volt-nuts@febo.com
> > > To unsubscribe, go to https://www.febo.com/cgi-bin/
> > > mailman/listinfo/volt-nuts
> > > and follow the instructions there.
> > >
> > _______________________________________________
> > volt-nuts mailing list -- volt-nuts@febo.com
> > To unsubscribe, go to https://www.febo.com/cgi-bin/
> > mailman/listinfo/volt-nuts
> > and follow the instructions there.
> >
> _______________________________________________
> volt-nuts mailing list -- volt-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/
> mailman/listinfo/volt-nuts
> and follow the instructions there.
>
RE
Randy Evans
Sun, Dec 31, 2017 5:54 AM
Todd,
Since the KE5FX NVRAM download software works, that would seem to indicate
the GPIB is working, wouldn't it (download completes and checksums OK)?
Randy
On Sat, Dec 30, 2017 at 9:52 PM, Randy Evans randyevans2688@gmail.com
wrote:
Todd,
The minimal quick test would be greatly appreciated if you have the time.
I know this is above and beyond what I would expect but it would be really
useful to see if there is any hope for the GPIB on this unit. I will do
the service note 13 procedure in the morning.
Thanks again,
Randy
On Sat, Dec 30, 2017 at 8:58 PM, Todd Micallef tmicallef@gmail.com
wrote:
Randy,
Short of swapping out the A5 digital board I have no other ideas. If you
don't have a NI or Agilent adapter, I am sure I can find a spare for you
to
test.
There is a Service Note 3458A-13A that deals with GPIB communication
failures. There is a test within the document to see if a shorting jumper
needs to be installed.
I can try reducing the code to something minimal for a quick test if you
want. It will take some time in the morning.
Todd
On Sat, Dec 30, 2017 at 11:18 PM, Randy Evans randyevans2688@gmail.com
wrote:
Todd,
Same as before. Still not working. All it does is print out "HP3458A"
the output console and then just sits there.
Randy
On Sat, Dec 30, 2017 at 7:43 PM, Todd Micallef tmicallef@gmail.com
wrote:
Randy,
I just remembered that I changed the duration from 30 minutes to 2
to speed up the test. I forgot to change it back.
Todd
On Sat, Dec 30, 2017 at 10:27 PM, Todd Micallef tmicallef@gmail.com
wrote:
Randy,
I just sent it. I haven't modified Martin's code yet. That can be
once you confirm it works for you.
Todd
On Sat, Dec 30, 2017 at 10:12 PM, randyevans2688 <
Todd,
If you don't mind, just send the code directly to me.
Randy
Sent from my Verizon, Samsung Galaxy smartphone
-------- Original message --------From: Todd Micallef <
tmicallef@gmail.com> Date: 12/30/17 6:40 PM (GMT-08:00) To:
of precise voltage measurement volt-nuts@febo.com Subject: Re:
[volt-nuts] GPIB Issue with 3458A
Randy,
I have not written anything else for the 3458A using EzGPIB. If
a
National Instruments or Agilent adapters, I have a handful of
I
use to measure some of my standards.
If you are reading the temperature from the HP meter, I would think
the problem lies in how the triggering is configured. The
been changed in this script so it may work for you. There was an
additional
command added for disabling the buffer.
The code can be posted here, if possible or I can email it to you
testing.
Todd
On Sat, Dec 30, 2017 at 9:26 PM, randyevans2688 <
Todd, thanks for the help. Where will you post the new code?
BTW, do you have any very simple code that is likely to work. I
interested in seeing if anything other than KE5FX's memory dump
will work on my HP3458A.
Thanks again,
Randy
Sent from my Verizon, Samsung Galaxy smartphone
-------- Original message --------From: Todd Micallef <
Date: 12/30/17 6:13 PM (GMT-08:00) To: Discussion of precise
3458A
Randy,
I found the Prologix adapters and have modified the code in the
script. I am doing a test run now and will post the script when
certain I didn't make it worse.
Todd
On Sat, Dec 30, 2017 at 7:15 PM, Randy Evans <
Todd,
I have swapped cables, moved the Prologix adapter to both
works, HP doesn't), tried Martins code as well as Xdevs's. I
tried another PC yet, maybe the next step.
Here is the code that Martin modified for two meters, as well
modified Xdev's code per your directions (V1.02).
Thanks,
Randy
On Sat, Dec 30, 2017 at 3:55 PM, Todd Micallef <
Randy,
I will continue to look for my adapter. I haven't used it for
while.
I have seen similar issues and swapped cables, moved the
other meter, etc... One of my cables was bad from one side of
to the other (on the same end of the cable) and caused me
time. Weird reading errors were happening.
Would you email me Martin's code so I can test it here and I
anything can be done to replicate your issue.
Todd
On Sat, Dec 30, 2017 at 6:42 PM, Randy Evans <
Todd,
I tried adding the line but it made no difference.
Randy
On Sat, Dec 30, 2017 at 12:19 PM, Randy Evans <
Todd,
Does it matter where I put the line in the Init_Device
procedure Init_Device;
begin;
EZGPIB_BusWriteData(gpib_address,GET_ID);
if EZGPIB_BusWaitForData(gpib_add
EZGPIB_ScreenWriteLn(meter_id);
EZGPIB_BusWriteData(gpib_address,TRIGGER_SOURCE);
until meter can be setup
EZGPIB_BusWriteData(gpib_address,'INBUF ON');
EZGPIB_BusWriteData(gpib_address,AUTO_ZERO);
EZGPIB_BusWriteData(gpib_address,OUTPUT_FORMAT);
end;
Randy
On Sat, Dec 30, 2017 at 11:20 AM, Todd Micallef <
Randy,
I am the person that wrote that EzGPIB script. It was my
moved on to Python and I have only written a couple for
meters I have on hand. Version 1.01 is the most current
the 3458A.
I see that I should have added at least one more
initialization. The additional init command is used in
xDevs.
Add the following line in the Init_Device procedure
EZGPIB_BusWriteData(gpib_address,'END ALWAYS');
This tells the meter to send an EOI on every read.
I am looking for one of my Prologix adapters or I would
find it, I will also test it. I want to make sure there
issues.
Also, do both meters run the same firmware versions? I
meter is using a much older version.
Todd
On Sat, Dec 30, 2017 at 10:42 AM, Randy Evans <
Todd,
I have version 1.01. Yes, please e-mail me your
will help.
Thanks,
Randy
On Sat, Dec 30, 2017 at 4:24 AM, Todd Micallef <
Randy,
Is there a version number at the top of the file? I
and 1.01). I can't remember all that changed between
been a while but I did find an additional timeout
wait time for reading data to 1 minute. I am
mine and added it.
I can email it to you if you want to try it. If it
have Illya update it on xDevs.
Todd
On Sat, Dec 30, 2017 at 12:39 AM, Randy Evans <
I am still unable to get my HP-3458A to work with
"HP3458A_Measure_DMM_Noise.488" (
It
works fine with my Agilent 3458A and with a
Agilent 3458A using the EZGPIB program. The SW
my friend's DMMs, the HP units are (8.2) and with
(9.2). However, both of my units work fine with
HP3458.EXE program to download the NVRAM CAL data.
indicate that my HP-3458 GPIB works at least
Does anyone have a EZGPIB program that I could
different from the Xdev program)? It would be
something to work other than the HP3458.EXE
some ideas on what to look for.
Thanks,
Randy Evans
On Tue, Dec 26, 2017 at 7:24 PM, Randy Evans <
I am hoping some of you experts might be able to
problem i am having trying to make automated
an Agilent and an HP 3458A. I have a program
Reynolds, which is a variant on the Xdevs noise
Martin's program continuously reads the voltage
from the two meters and then stores then in a
the temperature measurements from both meters.
his Agilent and HP 3458As.
On my two 3458As, it takes the temperature
takes a voltage measurement from the Agilent
make any voltage measurements although I can see
states and, of course, it does sent the
if I turn the HP unit off and then on again,
voltage and temperature measurements and send
Any ideas on what might be causing the strange
mailman/listinfo/volt-nuts
and follow the instructions there.
Todd,
Since the KE5FX NVRAM download software works, that would seem to indicate
the GPIB is working, wouldn't it (download completes and checksums OK)?
Randy
On Sat, Dec 30, 2017 at 9:52 PM, Randy Evans <randyevans2688@gmail.com>
wrote:
> Todd,
>
> The minimal quick test would be greatly appreciated if you have the time.
> I know this is above and beyond what I would expect but it would be really
> useful to see if there is any hope for the GPIB on this unit. I will do
> the service note 13 procedure in the morning.
>
> Thanks again,
>
> Randy
>
> On Sat, Dec 30, 2017 at 8:58 PM, Todd Micallef <tmicallef@gmail.com>
> wrote:
>
>> Randy,
>>
>> Short of swapping out the A5 digital board I have no other ideas. If you
>> don't have a NI or Agilent adapter, I am sure I can find a spare for you
>> to
>> test.
>>
>> There is a Service Note 3458A-13A that deals with GPIB communication
>> failures. There is a test within the document to see if a shorting jumper
>> needs to be installed.
>> I can try reducing the code to something minimal for a quick test if you
>> want. It will take some time in the morning.
>>
>> Todd
>>
>> On Sat, Dec 30, 2017 at 11:18 PM, Randy Evans <randyevans2688@gmail.com>
>> wrote:
>>
>> > Todd,
>> >
>> > Same as before. Still not working. All it does is print out "HP3458A"
>> on
>> > the output console and then just sits there.
>> >
>> > Randy
>> >
>> > On Sat, Dec 30, 2017 at 7:43 PM, Todd Micallef <tmicallef@gmail.com>
>> > wrote:
>> >
>> > > Randy,
>> > >
>> > > I just remembered that I changed the duration from 30 minutes to 2
>> > minutes
>> > > to speed up the test. I forgot to change it back.
>> > >
>> > > Todd
>> > >
>> > > On Sat, Dec 30, 2017 at 10:27 PM, Todd Micallef <tmicallef@gmail.com>
>> > > wrote:
>> > >
>> > > > Randy,
>> > > >
>> > > > I just sent it. I haven't modified Martin's code yet. That can be
>> done
>> > > > once you confirm it works for you.
>> > > >
>> > > > Todd
>> > > >
>> > > > On Sat, Dec 30, 2017 at 10:12 PM, randyevans2688 <
>> > > randyevans2688@gmail.com
>> > > > > wrote:
>> > > >
>> > > >> Todd,
>> > > >> If you don't mind, just send the code directly to me.
>> > > >> Randy
>> > > >>
>> > > >> Sent from my Verizon, Samsung Galaxy smartphone
>> > > >> -------- Original message --------From: Todd Micallef <
>> > > >> tmicallef@gmail.com> Date: 12/30/17 6:40 PM (GMT-08:00) To:
>> > > Discussion
>> > > >> of precise voltage measurement <volt-nuts@febo.com> Subject: Re:
>> > > >> [volt-nuts] GPIB Issue with 3458A
>> > > >> Randy,
>> > > >>
>> > > >> I have not written anything else for the 3458A using EzGPIB. If
>> you
>> > > have
>> > > >> a
>> > > >> National Instruments or Agilent adapters, I have a handful of
>> Python
>> > > code
>> > > >> I
>> > > >> use to measure some of my standards.
>> > > >>
>> > > >> If you are reading the temperature from the HP meter, I would think
>> > that
>> > > >> the problem lies in how the triggering is configured. The
>> triggering
>> > has
>> > > >> been changed in this script so it may work for you. There was an
>> > > >> additional
>> > > >> command added for disabling the buffer.
>> > > >> The code can be posted here, if possible or I can email it to you
>> for
>> > > >> testing.
>> > > >>
>> > > >> Todd
>> > > >>
>> > > >> On Sat, Dec 30, 2017 at 9:26 PM, randyevans2688 <
>> > > randyevans2688@gmail.com
>> > > >> >
>> > > >> wrote:
>> > > >>
>> > > >> > Todd, thanks for the help. Where will you post the new code?
>> > > >> > BTW, do you have any very simple code that is likely to work. I
>> am
>> > > >> > interested in seeing if anything other than KE5FX's memory dump
>> > > program
>> > > >> > will work on my HP3458A.
>> > > >> > Thanks again,
>> > > >> > Randy
>> > > >> >
>> > > >> > Sent from my Verizon, Samsung Galaxy smartphone
>> > > >> > -------- Original message --------From: Todd Micallef <
>> > > >> tmicallef@gmail.com>
>> > > >> > Date: 12/30/17 6:13 PM (GMT-08:00) To: Discussion of precise
>> > voltage
>> > > >> > measurement <volt-nuts@febo.com> Subject: Re: [volt-nuts] GPIB
>> > Issue
>> > > >> with
>> > > >> > 3458A
>> > > >> > Randy,
>> > > >> >
>> > > >> > I found the Prologix adapters and have modified the code in the
>> > > original
>> > > >> > script. I am doing a test run now and will post the script when
>> I am
>> > > >> > certain I didn't make it worse.
>> > > >> >
>> > > >> > Todd
>> > > >> >
>> > > >> > On Sat, Dec 30, 2017 at 7:15 PM, Randy Evans <
>> > > randyevans2688@gmail.com>
>> > > >> > wrote:
>> > > >> >
>> > > >> > > Todd,
>> > > >> > >
>> > > >> > > I have swapped cables, moved the Prologix adapter to both
>> meters
>> > > >> (Agilent
>> > > >> > > works, HP doesn't), tried Martins code as well as Xdevs's. I
>> have
>> > > not
>> > > >> > > tried another PC yet, maybe the next step.
>> > > >> > >
>> > > >> > > Here is the code that Martin modified for two meters, as well
>> as
>> > the
>> > > >> > > modified Xdev's code per your directions (V1.02).
>> > > >> > >
>> > > >> > > Thanks,
>> > > >> > >
>> > > >> > > Randy
>> > > >> > >
>> > > >> > > On Sat, Dec 30, 2017 at 3:55 PM, Todd Micallef <
>> > tmicallef@gmail.com
>> > > >
>> > > >> > > wrote:
>> > > >> > >
>> > > >> > > > Randy,
>> > > >> > > >
>> > > >> > > > I will continue to look for my adapter. I haven't used it for
>> > > quite
>> > > >> a
>> > > >> > > > while.
>> > > >> > > >
>> > > >> > > > I have seen similar issues and swapped cables, moved the
>> adapter
>> > > to
>> > > >> the
>> > > >> > > > other meter, etc... One of my cables was bad from one side of
>> > the
>> > > >> > > connector
>> > > >> > > > to the other (on the same end of the cable) and caused me
>> lots
>> > of
>> > > >> > wasted
>> > > >> > > > time. Weird reading errors were happening.
>> > > >> > > >
>> > > >> > > > Would you email me Martin's code so I can test it here and I
>> > will
>> > > >> see
>> > > >> > if
>> > > >> > > > anything can be done to replicate your issue.
>> > > >> > > >
>> > > >> > > > Todd
>> > > >> > > >
>> > > >> > > >
>> > > >> > > > On Sat, Dec 30, 2017 at 6:42 PM, Randy Evans <
>> > > >> randyevans2688@gmail.com
>> > > >> > >
>> > > >> > > > wrote:
>> > > >> > > >
>> > > >> > > > > Todd,
>> > > >> > > > >
>> > > >> > > > > I tried adding the line but it made no difference.
>> > > >> > > > >
>> > > >> > > > > Randy
>> > > >> > > > >
>> > > >> > > > > On Sat, Dec 30, 2017 at 12:19 PM, Randy Evans <
>> > > >> > > randyevans2688@gmail.com>
>> > > >> > > > > wrote:
>> > > >> > > > >
>> > > >> > > > > > Todd,
>> > > >> > > > > >
>> > > >> > > > > > Does it matter where I put the line in the Init_Device
>> > > >> procedure?
>> > > >> > > > > >
>> > > >> > > > > >
>> > > >> > > > > > procedure Init_Device;
>> > > >> > > > > > begin;
>> > > >> > > > > > EZGPIB_BusWriteData(gpib_address,GET_ID);
>> //
>> > > Get
>> > > >> ID
>> > > >> > of
>> > > >> > > > > meter
>> > > >> > > > > > if EZGPIB_BusWaitForData(gpib_add
>> ress,meter_id,Timeout)
>> > > then
>> > > >> > > > > > EZGPIB_ScreenWriteLn(meter_id);
>> > > >> > > > > >
>> > > >> > > > > > EZGPIB_BusWriteData(gpib_address,TRIGGER_SOURCE);
>> //
>> > > Hold
>> > > >> > > > trigger
>> > > >> > > > > > until meter can be setup
>> > > >> > > > > > EZGPIB_BusWriteData(gpib_address,'INBUF ON');
>> > > >> > > > > > EZGPIB_BusWriteData(gpib_address,AUTO_ZERO);
>> > > >> > > > > > EZGPIB_BusWriteData(gpib_address,OUTPUT_FORMAT);
>> > > >> > > > > > end;
>> > > >> > > > > >
>> > > >> > > > > >
>> > > >> > > > > > Randy
>> > > >> > > > > >
>> > > >> > > > > > On Sat, Dec 30, 2017 at 11:20 AM, Todd Micallef <
>> > > >> > tmicallef@gmail.com
>> > > >> > > >
>> > > >> > > > > > wrote:
>> > > >> > > > > >
>> > > >> > > > > >> Randy,
>> > > >> > > > > >>
>> > > >> > > > > >> I am the person that wrote that EzGPIB script. It was my
>> > > first
>> > > >> > > before
>> > > >> > > > I
>> > > >> > > > > >> moved on to Python and I have only written a couple for
>> the
>> > > >> > > different
>> > > >> > > > > >> meters I have on hand. Version 1.01 is the most current
>> > > >> version I
>> > > >> > > have
>> > > >> > > > > for
>> > > >> > > > > >> the 3458A.
>> > > >> > > > > >> I see that I should have added at least one more
>> command to
>> > > the
>> > > >> > > meter
>> > > >> > > > > >> initialization. The additional init command is used in
>> the
>> > > >> > scripts
>> > > >> > > on
>> > > >> > > > > >> xDevs.
>> > > >> > > > > >>
>> > > >> > > > > >> Add the following line in the Init_Device procedure
>> > > >> > > > > >> EZGPIB_BusWriteData(gpib_address,'END ALWAYS');
>> > > >> > > > > >>
>> > > >> > > > > >> This tells the meter to send an EOI on every read.
>> > > >> > > > > >>
>> > > >> > > > > >> I am looking for one of my Prologix adapters or I would
>> > test
>> > > >> this.
>> > > >> > > > When
>> > > >> > > > > I
>> > > >> > > > > >> find it, I will also test it. I want to make sure there
>> are
>> > > no
>> > > >> > > further
>> > > >> > > > > >> issues.
>> > > >> > > > > >> Also, do both meters run the same firmware versions? I
>> > wonder
>> > > >> if
>> > > >> > the
>> > > >> > > > HP
>> > > >> > > > > >> meter is using a much older version.
>> > > >> > > > > >>
>> > > >> > > > > >> Todd
>> > > >> > > > > >>
>> > > >> > > > > >> On Sat, Dec 30, 2017 at 10:42 AM, Randy Evans <
>> > > >> > > > randyevans2688@gmail.com
>> > > >> > > > > >
>> > > >> > > > > >> wrote:
>> > > >> > > > > >>
>> > > >> > > > > >> > Todd,
>> > > >> > > > > >> >
>> > > >> > > > > >> > I have version 1.01. Yes, please e-mail me your
>> modified
>> > > >> > version.
>> > > >> > > > > >> Maybe it
>> > > >> > > > > >> > will help.
>> > > >> > > > > >> >
>> > > >> > > > > >> > Thanks,
>> > > >> > > > > >> >
>> > > >> > > > > >> > Randy
>> > > >> > > > > >> >
>> > > >> > > > > >> > On Sat, Dec 30, 2017 at 4:24 AM, Todd Micallef <
>> > > >> > > tmicallef@gmail.com
>> > > >> > > > >
>> > > >> > > > > >> > wrote:
>> > > >> > > > > >> >
>> > > >> > > > > >> > > Randy,
>> > > >> > > > > >> > >
>> > > >> > > > > >> > > Is there a version number at the top of the file? I
>> had
>> > > two
>> > > >> > > > versions
>> > > >> > > > > >> > (1.00
>> > > >> > > > > >> > > and 1.01). I can't remember all that changed between
>> > the
>> > > >> two
>> > > >> > > since
>> > > >> > > > > it
>> > > >> > > > > >> has
>> > > >> > > > > >> > > been a while but I did find an additional timeout
>> code
>> > > that
>> > > >> > > > limited
>> > > >> > > > > >> the
>> > > >> > > > > >> > > wait time for reading data to 1 minute. I am
>> guessing I
>> > > >> had a
>> > > >> > > > > problem
>> > > >> > > > > >> > with
>> > > >> > > > > >> > > mine and added it.
>> > > >> > > > > >> > >
>> > > >> > > > > >> > > I can email it to you if you want to try it. If it
>> > works
>> > > >> for
>> > > >> > > you,
>> > > >> > > > I
>> > > >> > > > > >> will
>> > > >> > > > > >> > > have Illya update it on xDevs.
>> > > >> > > > > >> > >
>> > > >> > > > > >> > > Todd
>> > > >> > > > > >> > >
>> > > >> > > > > >> > > On Sat, Dec 30, 2017 at 12:39 AM, Randy Evans <
>> > > >> > > > > >> randyevans2688@gmail.com>
>> > > >> > > > > >> > > wrote:
>> > > >> > > > > >> > >
>> > > >> > > > > >> > > > I am still unable to get my HP-3458A to work with
>> the
>> > > >> xdevs
>> > > >> > > > > program
>> > > >> > > > > >> > > > "HP3458A_Measure_DMM_Noise.488" (
>> > > >> > > https://xdevs.com/article/dmm
>> > > >> > > > > >> _noise/
>> > > >> > > > > >> > ).
>> > > >> > > > > >> > > > It
>> > > >> > > > > >> > > > works fine with my Agilent 3458A and with a
>> friend's
>> > > >> > HP-3458A
>> > > >> > > > and
>> > > >> > > > > >> his
>> > > >> > > > > >> > > > Agilent 3458A using the EZGPIB program. The SW
>> > > versions
>> > > >> are
>> > > >> > > > > >> identical
>> > > >> > > > > >> > > with
>> > > >> > > > > >> > > > my friend's DMMs, the HP units are (8.2) and with
>> the
>> > > >> > Agilent
>> > > >> > > > > units
>> > > >> > > > > >> are
>> > > >> > > > > >> > > > (9.2). However, both of my units work fine with
>> the
>> > > >> KE5FX
>> > > >> > > GPIB
>> > > >> > > > > >> Toolkit
>> > > >> > > > > >> > > > HP3458.EXE program to download the NVRAM CAL data.
>> > > This
>> > > >> > would
>> > > >> > > > > seem
>> > > >> > > > > >> to
>> > > >> > > > > >> > > > indicate that my HP-3458 GPIB works at least
>> > partially.
>> > > >> > > > > >> > > >
>> > > >> > > > > >> > > > Does anyone have a EZGPIB program that I could
>> try to
>> > > run
>> > > >> > > (that
>> > > >> > > > is
>> > > >> > > > > >> > > > different from the Xdev program)? It would be
>> nice
>> > to
>> > > >> see
>> > > >> > if
>> > > >> > > I
>> > > >> > > > > >> could
>> > > >> > > > > >> > get
>> > > >> > > > > >> > > > something to work other than the HP3458.EXE
>> program.
>> > > Or
>> > > >> if
>> > > >> > > any
>> > > >> > > > > one
>> > > >> > > > > >> has
>> > > >> > > > > >> > > > some ideas on what to look for.
>> > > >> > > > > >> > > >
>> > > >> > > > > >> > > > Thanks,
>> > > >> > > > > >> > > >
>> > > >> > > > > >> > > > Randy Evans
>> > > >> > > > > >> > > >
>> > > >> > > > > >> > > > On Tue, Dec 26, 2017 at 7:24 PM, Randy Evans <
>> > > >> > > > > >> randyevans2688@gmail.com
>> > > >> > > > > >> > >
>> > > >> > > > > >> > > > wrote:
>> > > >> > > > > >> > > >
>> > > >> > > > > >> > > > > I am hoping some of you experts might be able to
>> > > assist
>> > > >> > > with a
>> > > >> > > > > >> > strange
>> > > >> > > > > >> > > > > problem i am having trying to make automated
>> > > >> measurements
>> > > >> > > > using
>> > > >> > > > > >> > EZGPIB
>> > > >> > > > > >> > > > with
>> > > >> > > > > >> > > > > an Agilent and an HP 3458A. I have a program
>> given
>> > > to
>> > > >> me
>> > > >> > by
>> > > >> > > > > >> Martin
>> > > >> > > > > >> > > > > Reynolds, which is a variant on the Xdevs noise
>> > > >> > measurement
>> > > >> > > > > >> program.
>> > > >> > > > > >> > > > > Martin's program continuously reads the voltage
>> > > >> > measurements
>> > > >> > > > > >> > > sequentially
>> > > >> > > > > >> > > > > from the two meters and then stores then in a
>> > > >> measurements
>> > > >> > > > file,
>> > > >> > > > > >> > along
>> > > >> > > > > >> > > > with
>> > > >> > > > > >> > > > > the temperature measurements from both meters.
>> The
>> > > >> > program
>> > > >> > > > > works
>> > > >> > > > > >> > fine
>> > > >> > > > > >> > > on
>> > > >> > > > > >> > > > > his Agilent and HP 3458As.
>> > > >> > > > > >> > > > >
>> > > >> > > > > >> > > > > On my two 3458As, it takes the temperature
>> > > measurements
>> > > >> > just
>> > > >> > > > > fine
>> > > >> > > > > >> but
>> > > >> > > > > >> > > > only
>> > > >> > > > > >> > > > > takes a voltage measurement from the Agilent
>> unit.
>> > > >> The HP
>> > > >> > > > unit
>> > > >> > > > > >> will
>> > > >> > > > > >> > > not
>> > > >> > > > > >> > > > > make any voltage measurements although I can see
>> > both
>> > > >> the
>> > > >> > > > 3458As
>> > > >> > > > > >> in
>> > > >> > > > > >> > > talk
>> > > >> > > > > >> > > > > states and, of course, it does sent the
>> temperature
>> > > >> > > > > measurements.
>> > > >> > > > > >> > > > However,
>> > > >> > > > > >> > > > > if I turn the HP unit off and then on again,
>> the HP
>> > > >> will
>> > > >> > > start
>> > > >> > > > > to
>> > > >> > > > > >> > take
>> > > >> > > > > >> > > > > voltage and temperature measurements and send
>> them
>> > to
>> > > >> the
>> > > >> > > PC.
>> > > >> > > > > >> > > > >
>> > > >> > > > > >> > > > > Any ideas on what might be causing the strange
>> > > >> behavior?
>> > > >> > > > > >> > > > >
>> > > >> > > > > >> > > > > Thanks,
>> > > >> > > > > >> > > > >
>> > > >> > > > > >> > > > > Randy Evans
>> > > >> > > > > >> > > > >
>> > > >> > > > > >> > > > _______________________________________________
>> > > >> > > > > >> > > > volt-nuts mailing list -- volt-nuts@febo.com
>> > > >> > > > > >> > > > To unsubscribe, go to
>> https://www.febo.com/cgi-bin/
>> > > >> > > > > >> > > > mailman/listinfo/volt-nuts
>> > > >> > > > > >> > > > and follow the instructions there.
>> > > >> > > > > >> > > >
>> > > >> > > > > >> > > _______________________________________________
>> > > >> > > > > >> > > volt-nuts mailing list -- volt-nuts@febo.com
>> > > >> > > > > >> > > To unsubscribe, go to https://www.febo.com/cgi-bin/
>> > > >> > > > > >> > > mailman/listinfo/volt-nuts
>> > > >> > > > > >> > > and follow the instructions there.
>> > > >> > > > > >> > >
>> > > >> > > > > >> > _______________________________________________
>> > > >> > > > > >> > volt-nuts mailing list -- volt-nuts@febo.com
>> > > >> > > > > >> > To unsubscribe, go to https://www.febo.com/cgi-bin/
>> > > >> > > > > >> > mailman/listinfo/volt-nuts
>> > > >> > > > > >> > and follow the instructions there.
>> > > >> > > > > >> >
>> > > >> > > > > >> _______________________________________________
>> > > >> > > > > >> volt-nuts mailing list -- volt-nuts@febo.com
>> > > >> > > > > >> To unsubscribe, go to https://www.febo.com/cgi-bin/m
>> > > >> > > > > >> ailman/listinfo/volt-nuts
>> > > >> > > > > >> and follow the instructions there.
>> > > >> > > > > >>
>> > > >> > > > > >
>> > > >> > > > > >
>> > > >> > > > > _______________________________________________
>> > > >> > > > > volt-nuts mailing list -- volt-nuts@febo.com
>> > > >> > > > > To unsubscribe, go to https://www.febo.com/cgi-bin/
>> > > >> > > > > mailman/listinfo/volt-nuts
>> > > >> > > > > and follow the instructions there.
>> > > >> > > > >
>> > > >> > > > _______________________________________________
>> > > >> > > > volt-nuts mailing list -- volt-nuts@febo.com
>> > > >> > > > To unsubscribe, go to https://www.febo.com/cgi-bin/
>> > > >> > > > mailman/listinfo/volt-nuts
>> > > >> > > > and follow the instructions there.
>> > > >> > > >
>> > > >> > >
>> > > >> > > _______________________________________________
>> > > >> > > volt-nuts mailing list -- volt-nuts@febo.com
>> > > >> > > To unsubscribe, go to https://www.febo.com/cgi-bin/
>> > > >> > > mailman/listinfo/volt-nuts
>> > > >> > > and follow the instructions there.
>> > > >> > >
>> > > >> > _______________________________________________
>> > > >> > volt-nuts mailing list -- volt-nuts@febo.com
>> > > >> > To unsubscribe, go to https://www.febo.com/cgi-bin/
>> > > >> > mailman/listinfo/volt-nuts
>> > > >> > and follow the instructions there.
>> > > >> > _______________________________________________
>> > > >> > volt-nuts mailing list -- volt-nuts@febo.com
>> > > >> > To unsubscribe, go to https://www.febo.com/cgi-bin/
>> > > >> > mailman/listinfo/volt-nuts
>> > > >> > and follow the instructions there.
>> > > >> >
>> > > >> _______________________________________________
>> > > >> volt-nuts mailing list -- volt-nuts@febo.com
>> > > >> To unsubscribe, go to https://www.febo.com/cgi-bin/m
>> > > >> ailman/listinfo/volt-nuts
>> > > >> and follow the instructions there.
>> > > >> _______________________________________________
>> > > >> volt-nuts mailing list -- volt-nuts@febo.com
>> > > >> To unsubscribe, go to https://www.febo.com/cgi-bin/m
>> > > >> ailman/listinfo/volt-nuts
>> > > >> and follow the instructions there.
>> > > >>
>> > > >
>> > > >
>> > > _______________________________________________
>> > > volt-nuts mailing list -- volt-nuts@febo.com
>> > > To unsubscribe, go to https://www.febo.com/cgi-bin/
>> > > mailman/listinfo/volt-nuts
>> > > and follow the instructions there.
>> > >
>> > _______________________________________________
>> > volt-nuts mailing list -- volt-nuts@febo.com
>> > To unsubscribe, go to https://www.febo.com/cgi-bin/
>> > mailman/listinfo/volt-nuts
>> > and follow the instructions there.
>> >
>> _______________________________________________
>> volt-nuts mailing list -- volt-nuts@febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/m
>> ailman/listinfo/volt-nuts
>> and follow the instructions there.
>>
>
>
PK
Poul-Henning Kamp
Sun, Dec 31, 2017 8:31 AM
Since the KE5FX NVRAM download software works, that would seem to indicate
the GPIB is working, wouldn't it (download completes and checksums OK)?
Randy,
I think you need to consider electrical issues now.
GPIB buffer chips live a tough life, and they can fail or go frail
in so many weird ways that it is not even funny any longer.
The bad thing is that the problem may not even be the buffer chip
in the instrument you are talking to :-(
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
--------
In message <CANwu9JbP61knpuXEOzyGS8UmR2KuViA+zb4jDSYpsTw12OqS0Q@mail.gmail.com>
, Randy Evans writes:
>Since the KE5FX NVRAM download software works, that would seem to indicate
>the GPIB is working, wouldn't it (download completes and checksums OK)?
Randy,
I think you need to consider electrical issues now.
GPIB buffer chips live a tough life, and they can fail or go frail
in so many weird ways that it is not even funny any longer.
The bad thing is that the problem may not even be the buffer chip
in the instrument you are talking to :-(
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
TM
Todd Micallef
Sun, Dec 31, 2017 4:33 PM
In message <CANwu9JbP61knpuXEOzyGS8UmR2KuViA+zb4jDSYpsTw12OqS0Q@mail.
gmail.com>
, Randy Evans writes:
Since the KE5FX NVRAM download software works, that would seem to indicate
the GPIB is working, wouldn't it (download completes and checksums OK)?
Randy,
I think you need to consider electrical issues now.
GPIB buffer chips live a tough life, and they can fail or go frail
in so many weird ways that it is not even funny any longer.
The bad thing is that the problem may not even be the buffer chip
in the instrument you are talking to :-(
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk@FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/volt-nuts
and follow the instructions there.
Randy,
Check that your firmware version on the Prologix is up to date. The latest
USB version is 6.107. You can see it in the Prologix GPIB Configurator
utility that can be downloaded from a link on their website.
http://prologix.biz/resources.html
I have uploaded a pic of my settings shown in the utility.
https://pasteboard.co/H0MnDT1.png
Todd
On Sun, Dec 31, 2017 at 3:31 AM, Poul-Henning Kamp <phk@phk.freebsd.dk>
wrote:
> --------
> In message <CANwu9JbP61knpuXEOzyGS8UmR2KuViA+zb4jDSYpsTw12OqS0Q@mail.
> gmail.com>
> , Randy Evans writes:
>
> >Since the KE5FX NVRAM download software works, that would seem to indicate
> >the GPIB is working, wouldn't it (download completes and checksums OK)?
>
> Randy,
>
> I think you need to consider electrical issues now.
>
> GPIB buffer chips live a tough life, and they can fail or go frail
> in so many weird ways that it is not even funny any longer.
>
> The bad thing is that the problem may not even be the buffer chip
> in the instrument you are talking to :-(
>
>
> --
> Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
> phk@FreeBSD.ORG | TCP/IP since RFC 956
> FreeBSD committer | BSD since 4.3-tahoe
> Never attribute to malice what can adequately be explained by incompetence.