volt-nuts@lists.febo.com

Discussion of precise voltage measurement

View all threads

HP3458 ADC integrator

AK
Attila Kinali
Fri, Feb 23, 2018 3:51 PM

Moin,

I am trying to understand how the HP 3458 integrator of its
dual-slope ADC works, as it's a bit more complicated than
a simple, single opamp integrator. As you can see from the
attached picture, it consists of 3 opamps (U1-U3) for the
integrator itself, and another for the slope amplification (U4).
Unfortunately, I do not get what U2 and U3 are doing, much less
why they are needed.

So I tried to simulate it in LTspice. But, the circuit oscillates.
In both, the reset state (when the integration-C shorting switch is closed)
and the integration state (though, less wildly so). I guess, this
is because I replaced the LM6361 and AD848 by the LT1225 but do not
understand why, or respectively, which of the differing parameters of
the LT1225 causes this oscillations.

Would someone be so kind and give me some hints?

Thanks in advance

		Attila Kinali

--
It is upon moral qualities that a society is ultimately founded. All
the prosperity and technological sophistication in the world is of no
use without that foundation.
-- Miss Matheson, The Diamond Age, Neil Stephenson

Moin, I am trying to understand how the HP 3458 integrator of its dual-slope ADC works, as it's a bit more complicated than a simple, single opamp integrator. As you can see from the attached picture, it consists of 3 opamps (U1-U3) for the integrator itself, and another for the slope amplification (U4). Unfortunately, I do not get what U2 and U3 are doing, much less why they are needed. So I tried to simulate it in LTspice. But, the circuit oscillates. In both, the reset state (when the integration-C shorting switch is closed) and the integration state (though, less wildly so). I guess, this is because I replaced the LM6361 and AD848 by the LT1225 but do not understand why, or respectively, which of the differing parameters of the LT1225 causes this oscillations. Would someone be so kind and give me some hints? Thanks in advance Attila Kinali -- It is upon moral qualities that a society is ultimately founded. All the prosperity and technological sophistication in the world is of no use without that foundation. -- Miss Matheson, The Diamond Age, Neil Stephenson
DC
David C. Partridge
Fri, Feb 23, 2018 5:14 PM

The April 1989 HP Journal is mostly dedicated to the 3458A and there's an
article about the dual slope integrator starting on page 8.

There's also a nice discussion about it in Art of Electronics 3rd Edition
starting on page 918

HtH
David
-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Attila
Kinali
Sent: 23 February 2018 15:51
To: Discussion of precise voltage measurement
Subject: [volt-nuts] HP3458 ADC integrator

Moin,

I am trying to understand how the HP 3458 integrator of its dual-slope ADC
works, as it's a bit more complicated than a simple, single opamp
integrator. As you can see from the attached picture, it consists of 3
opamps (U1-U3) for the integrator itself, and another for the slope
amplification (U4).
Unfortunately, I do not get what U2 and U3 are doing, much less why they are
needed.

So I tried to simulate it in LTspice. But, the circuit oscillates.
In both, the reset state (when the integration-C shorting switch is closed)
and the integration state (though, less wildly so). I guess, this is because
I replaced the LM6361 and AD848 by the LT1225 but do not understand why, or
respectively, which of the differing parameters of the LT1225 causes this
oscillations.

Would someone be so kind and give me some hints?

Thanks in advance

		Attila Kinali

--
It is upon moral qualities that a society is ultimately founded. All the
prosperity and technological sophistication in the world is of no use
without that foundation.
-- Miss Matheson, The Diamond Age, Neil Stephenson

The April 1989 HP Journal is mostly dedicated to the 3458A and there's an article about the dual slope integrator starting on page 8. There's also a nice discussion about it in Art of Electronics 3rd Edition starting on page 918 HtH David -----Original Message----- From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Attila Kinali Sent: 23 February 2018 15:51 To: Discussion of precise voltage measurement Subject: [volt-nuts] HP3458 ADC integrator Moin, I am trying to understand how the HP 3458 integrator of its dual-slope ADC works, as it's a bit more complicated than a simple, single opamp integrator. As you can see from the attached picture, it consists of 3 opamps (U1-U3) for the integrator itself, and another for the slope amplification (U4). Unfortunately, I do not get what U2 and U3 are doing, much less why they are needed. So I tried to simulate it in LTspice. But, the circuit oscillates. In both, the reset state (when the integration-C shorting switch is closed) and the integration state (though, less wildly so). I guess, this is because I replaced the LM6361 and AD848 by the LT1225 but do not understand why, or respectively, which of the differing parameters of the LT1225 causes this oscillations. Would someone be so kind and give me some hints? Thanks in advance Attila Kinali -- It is upon moral qualities that a society is ultimately founded. All the prosperity and technological sophistication in the world is of no use without that foundation. -- Miss Matheson, The Diamond Age, Neil Stephenson
AK
Attila Kinali
Fri, Feb 23, 2018 5:31 PM

On Fri, 23 Feb 2018 17:14:28 -0000
"David C. Partridge" david.partridge@perdrix.co.uk wrote:

The April 1989 HP Journal is mostly dedicated to the 3458A and there's an
article about the dual slope integrator starting on page 8.

There's also a nice discussion about it in Art of Electronics 3rd Edition
starting on page 918

The HP journal does only mentions a single opamp integrator. The Art
of electronics only deals with the close relative used in the HP 34420,
which is similar, but not the same. It uses a two opamp structure,
which TAoE3 explains to be to increase amplification and slewrate.
I guess that can be said about the HP3458 as well, but that begs the
question why did they chose a 3 opamp structure? And why are U2 and U3
wired up the way they are?

			Attila Kinali

--
It is upon moral qualities that a society is ultimately founded. All
the prosperity and technological sophistication in the world is of no
use without that foundation.
-- Miss Matheson, The Diamond Age, Neil Stephenson

On Fri, 23 Feb 2018 17:14:28 -0000 "David C. Partridge" <david.partridge@perdrix.co.uk> wrote: > The April 1989 HP Journal is mostly dedicated to the 3458A and there's an > article about the dual slope integrator starting on page 8. > > There's also a nice discussion about it in Art of Electronics 3rd Edition > starting on page 918 The HP journal does only mentions a single opamp integrator. The Art of electronics only deals with the close relative used in the HP 34420, which is similar, but not the same. It uses a two opamp structure, which TAoE3 explains to be to increase amplification and slewrate. I guess that can be said about the HP3458 as well, but that begs the question why did they chose a 3 opamp structure? And why are U2 and U3 wired up the way they are? Attila Kinali -- It is upon moral qualities that a society is ultimately founded. All the prosperity and technological sophistication in the world is of no use without that foundation. -- Miss Matheson, The Diamond Age, Neil Stephenson
AK
Attila Kinali
Fri, Feb 23, 2018 6:01 PM

On Fri, 23 Feb 2018 18:31:47 +0100
Attila Kinali attila@kinali.ch wrote:

And why are U2 and U3 wired up the way they are?

Let me expand here a little bit to make it clear what puzzles me.
The HP3458 circuit uses 3 opamps: LT1001, LM6361 and AD848.
The LT1001 is there for its low input bias current of typ 1.2ns,
but has a GBW of typ 0.8MHz, SR of 0.25Vµs and A_vol of 700 V/mV.
The LM6361 comes with GBW of 40MHz, SR of 200 V/µs and A_vol of 0.55 V/mV.
The AD848 has a GBW of 175MHz, SR of 200 V/µs and A_vol of 13 V/mV.

Just by judging from the slew rate, the LM6361 alone should be enough
to get the ciruit up to speed, and the AD848 wouldnt be needed.
The AD848 has a much higher GBW and A_vol and I can see that
both would be beneficial for the integrator. But why not just use
a AD848 and leave the LM6361 out? Why use both?

The jfet is obviously there to decouple the high speed opamps with
their high bias current inputs from the sensitive summing point.
C2 is probably to limit the slew rate of U2 in order to keep the
integrator stable. But this means that U2+U3 together have some
stability problem that could be solved by removing one (preferably
U2 as it's the slower of the two).

		Attila Kinali

--
It is upon moral qualities that a society is ultimately founded. All
the prosperity and technological sophistication in the world is of no
use without that foundation.
-- Miss Matheson, The Diamond Age, Neil Stephenson

On Fri, 23 Feb 2018 18:31:47 +0100 Attila Kinali <attila@kinali.ch> wrote: > And why are U2 and U3 wired up the way they are? Let me expand here a little bit to make it clear what puzzles me. The HP3458 circuit uses 3 opamps: LT1001, LM6361 and AD848. The LT1001 is there for its low input bias current of typ 1.2ns, but has a GBW of typ 0.8MHz, SR of 0.25Vµs and A_vol of 700 V/mV. The LM6361 comes with GBW of 40MHz, SR of 200 V/µs and A_vol of 0.55 V/mV. The AD848 has a GBW of 175MHz, SR of 200 V/µs and A_vol of 13 V/mV. Just by judging from the slew rate, the LM6361 alone should be enough to get the ciruit up to speed, and the AD848 wouldnt be needed. The AD848 has a much higher GBW and A_vol and I can see that both would be beneficial for the integrator. But why not just use a AD848 and leave the LM6361 out? Why use both? The jfet is obviously there to decouple the high speed opamps with their high bias current inputs from the sensitive summing point. C2 is probably to limit the slew rate of U2 in order to keep the integrator stable. But this means that U2+U3 together have some stability problem that could be solved by removing one (preferably U2 as it's the slower of the two). Attila Kinali -- It is upon moral qualities that a society is ultimately founded. All the prosperity and technological sophistication in the world is of no use without that foundation. -- Miss Matheson, The Diamond Age, Neil Stephenson
DC
David C. Partridge
Fri, Feb 23, 2018 6:48 PM

Looking your schematic, I see that you are using LT1225 for both U2 (U111) and U3 (U112)

Why did you use those rather than LM6361 and AD848?  Was it just because LTSpice didn't have models for them?  It can make a difference ...

As for why they chose to use 3 op-amps rather than 2?  I don't know except that when dealing with low level signals maybe they needed the extra gain.

BTW LM6361 is 550V/mv not .55V (I assume that was a mistake in your post) ... Slew rates seem a bit different at +/- 15V supply as well.

David

-----Original Message-----
From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Attila Kinali
Sent: 23 February 2018 18:02
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] HP3458 ADC integrator

On Fri, 23 Feb 2018 18:31:47 +0100
Attila Kinali attila@kinali.ch wrote:

And why are U2 and U3 wired up the way they are?

Let me expand here a little bit to make it clear what puzzles me.
The HP3458 circuit uses 3 opamps: LT1001, LM6361 and AD848.
The LT1001 is there for its low input bias current of typ 1.2ns, but has a GBW of typ 0.8MHz, SR of 0.25Vµs and A_vol of 700 V/mV.
The LM6361 comes with GBW of 40MHz, SR of 200 V/µs and A_vol of 0.55 V/mV.
The AD848 has a GBW of 175MHz, SR of 200 V/µs and A_vol of 13 V/mV.

Just by judging from the slew rate, the LM6361 alone should be enough to get the ciruit up to speed, and the AD848 wouldnt be needed.
The AD848 has a much higher GBW and A_vol and I can see that both would be beneficial for the integrator. But why not just use a AD848 and leave the LM6361 out? Why use both?

The jfet is obviously there to decouple the high speed opamps with their high bias current inputs from the sensitive summing point.
C2 is probably to limit the slew rate of U2 in order to keep the integrator stable. But this means that U2+U3 together have some stability problem that could be solved by removing one (preferably
U2 as it's the slower of the two).

		Attila Kinali

--
It is upon moral qualities that a society is ultimately founded. All the prosperity and technological sophistication in the world is of no use without that foundation.
-- Miss Matheson, The Diamond Age, Neil Stephenson _______________________________________________
volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Looking your schematic, I see that you are using LT1225 for both U2 (U111) and U3 (U112) Why did you use those rather than LM6361 and AD848? Was it just because LTSpice didn't have models for them? It can make a difference ... As for why they chose to use 3 op-amps rather than 2? I don't know except that when dealing with low level signals maybe they needed the extra gain. BTW LM6361 is 550V/mv not .55V (I assume that was a mistake in your post) ... Slew rates seem a bit different at +/- 15V supply as well. David -----Original Message----- From: volt-nuts [mailto:volt-nuts-bounces@febo.com] On Behalf Of Attila Kinali Sent: 23 February 2018 18:02 To: Discussion of precise voltage measurement Subject: Re: [volt-nuts] HP3458 ADC integrator On Fri, 23 Feb 2018 18:31:47 +0100 Attila Kinali <attila@kinali.ch> wrote: > And why are U2 and U3 wired up the way they are? Let me expand here a little bit to make it clear what puzzles me. The HP3458 circuit uses 3 opamps: LT1001, LM6361 and AD848. The LT1001 is there for its low input bias current of typ 1.2ns, but has a GBW of typ 0.8MHz, SR of 0.25Vµs and A_vol of 700 V/mV. The LM6361 comes with GBW of 40MHz, SR of 200 V/µs and A_vol of 0.55 V/mV. The AD848 has a GBW of 175MHz, SR of 200 V/µs and A_vol of 13 V/mV. Just by judging from the slew rate, the LM6361 alone should be enough to get the ciruit up to speed, and the AD848 wouldnt be needed. The AD848 has a much higher GBW and A_vol and I can see that both would be beneficial for the integrator. But why not just use a AD848 and leave the LM6361 out? Why use both? The jfet is obviously there to decouple the high speed opamps with their high bias current inputs from the sensitive summing point. C2 is probably to limit the slew rate of U2 in order to keep the integrator stable. But this means that U2+U3 together have some stability problem that could be solved by removing one (preferably U2 as it's the slower of the two). Attila Kinali -- It is upon moral qualities that a society is ultimately founded. All the prosperity and technological sophistication in the world is of no use without that foundation. -- Miss Matheson, The Diamond Age, Neil Stephenson _______________________________________________ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
BG
Bruce Griffiths
Fri, Feb 23, 2018 7:45 PM

Moin attila

I'd start with simulating and understanding the integrator in the 34401A, its a little simpler but still has a coupe of opamps in the integrator. Then analysing the GBW requirements of the various opamps used in the HP3458A shoul become somewhat clearer.

Bruce

 On 24 February 2018 at 04:51 Attila Kinali <attila@kinali.ch> wrote:

 Moin,

 I am trying to understand how the HP 3458 integrator of its
 dual-slope ADC works, as it's a bit more complicated than
 a simple, single opamp integrator. As you can see from the
 attached picture, it consists of 3 opamps (U1-U3) for the
 integrator itself, and another for the slope amplification (U4).
 Unfortunately, I do not get what U2 and U3 are doing, much less
 why they are needed.

 So I tried to simulate it in LTspice. But, the circuit oscillates.
 In both, the reset state (when the integration-C shorting switch is closed)
 and the integration state (though, less wildly so). I guess, this
 is because I replaced the LM6361 and AD848 by the LT1225 but do not
 understand why, or respectively, which of the differing parameters of
 the LT1225 causes this oscillations.

 Would someone be so kind and give me some hints?

 Thanks in advance

 Attila Kinali

 --
 It is upon moral qualities that a society is ultimately founded. All
 the prosperity and technological sophistication in the world is of no
 use without that foundation.
 -- Miss Matheson, The Diamond Age, Neil Stephenson

 _______________________________________________
 volt-nuts mailing list -- volt-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
 and follow the instructions there.
Moin attila I'd start with simulating and understanding the integrator in the 34401A, its a little simpler but still has a coupe of opamps in the integrator. Then analysing the GBW requirements of the various opamps used in the HP3458A shoul become somewhat clearer. Bruce > > On 24 February 2018 at 04:51 Attila Kinali <attila@kinali.ch> wrote: > > Moin, > > I am trying to understand how the HP 3458 integrator of its > dual-slope ADC works, as it's a bit more complicated than > a simple, single opamp integrator. As you can see from the > attached picture, it consists of 3 opamps (U1-U3) for the > integrator itself, and another for the slope amplification (U4). > Unfortunately, I do not get what U2 and U3 are doing, much less > why they are needed. > > So I tried to simulate it in LTspice. But, the circuit oscillates. > In both, the reset state (when the integration-C shorting switch is closed) > and the integration state (though, less wildly so). I guess, this > is because I replaced the LM6361 and AD848 by the LT1225 but do not > understand why, or respectively, which of the differing parameters of > the LT1225 causes this oscillations. > > Would someone be so kind and give me some hints? > > Thanks in advance > > Attila Kinali > > -- > It is upon moral qualities that a society is ultimately founded. All > the prosperity and technological sophistication in the world is of no > use without that foundation. > -- Miss Matheson, The Diamond Age, Neil Stephenson > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. >
BG
Bruce Griffiths
Fri, Feb 23, 2018 8:49 PM

Hint: start by looking at the summing junction voltage of a single opamp integrator with R = 50k, C = 330pF.

NB the LT1225 isnt unity gain stable so its use for any of U2, U3 and U4 is questionable.

They will all likely oscillate on their own.

Bruce

 On 24 February 2018 at 08:45 Bruce Griffiths <bruce.griffiths@xtra.co.nz> wrote:

 Moin attila

 I'd start with simulating and understanding the integrator in the 34401A, its a little simpler but still has a coupe of opamps in the integrator. Then analysing the GBW requirements of the various opamps used in the HP3458A shoul become somewhat clearer.

 Bruce
     On 24 February 2018 at 04:51 Attila Kinali <attila@kinali.ch> wrote:

     Moin,

     I am trying to understand how the HP 3458 integrator of its
     dual-slope ADC works, as it's a bit more complicated than
     a simple, single opamp integrator. As you can see from the
     attached picture, it consists of 3 opamps (U1-U3) for the
     integrator itself, and another for the slope amplification (U4).
     Unfortunately, I do not get what U2 and U3 are doing, much less
     why they are needed.

     So I tried to simulate it in LTspice. But, the circuit oscillates.
     In both, the reset state (when the integration-C shorting switch is closed)
     and the integration state (though, less wildly so). I guess, this
     is because I replaced the LM6361 and AD848 by the LT1225 but do not
     understand why, or respectively, which of the differing parameters of
     the LT1225 causes this oscillations.

     Would someone be so kind and give me some hints?

     Thanks in advance

     Attila Kinali

     --
     It is upon moral qualities that a society is ultimately founded. All
     the prosperity and technological sophistication in the world is of no
     use without that foundation.
     -- Miss Matheson, The Diamond Age, Neil Stephenson

     _______________________________________________
     volt-nuts mailing list -- volt-nuts@febo.com
     To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
     and follow the instructions there.

     _______________________________________________
     volt-nuts mailing list -- volt-nuts@febo.com
     To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
     and follow the instructions there.
Hint: start by looking at the summing junction voltage of a single opamp integrator with R = 50k, C = 330pF. NB the LT1225 isnt unity gain stable so its use for any of U2, U3 and U4 is questionable. They will all likely oscillate on their own. Bruce > > On 24 February 2018 at 08:45 Bruce Griffiths <bruce.griffiths@xtra.co.nz> wrote: > > Moin attila > > I'd start with simulating and understanding the integrator in the 34401A, its a little simpler but still has a coupe of opamps in the integrator. Then analysing the GBW requirements of the various opamps used in the HP3458A shoul become somewhat clearer. > > Bruce > > > > > > On 24 February 2018 at 04:51 Attila Kinali <attila@kinali.ch> wrote: > > > > Moin, > > > > I am trying to understand how the HP 3458 integrator of its > > dual-slope ADC works, as it's a bit more complicated than > > a simple, single opamp integrator. As you can see from the > > attached picture, it consists of 3 opamps (U1-U3) for the > > integrator itself, and another for the slope amplification (U4). > > Unfortunately, I do not get what U2 and U3 are doing, much less > > why they are needed. > > > > So I tried to simulate it in LTspice. But, the circuit oscillates. > > In both, the reset state (when the integration-C shorting switch is closed) > > and the integration state (though, less wildly so). I guess, this > > is because I replaced the LM6361 and AD848 by the LT1225 but do not > > understand why, or respectively, which of the differing parameters of > > the LT1225 causes this oscillations. > > > > Would someone be so kind and give me some hints? > > > > Thanks in advance > > > > Attila Kinali > > > > -- > > It is upon moral qualities that a society is ultimately founded. All > > the prosperity and technological sophistication in the world is of no > > use without that foundation. > > -- Miss Matheson, The Diamond Age, Neil Stephenson > > > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > and follow the instructions there. > > > > _______________________________________________ > > volt-nuts mailing list -- volt-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > > and follow the instructions there. > > > > >
AK
Attila Kinali
Fri, Feb 23, 2018 9:23 PM

On Fri, 23 Feb 2018 18:48:24 -0000
"David C. Partridge" david.partridge@perdrix.co.uk wrote:

Looking your schematic, I see that you are using LT1225 for both U2 (U111) and U3 (U112)

Why did you use those rather than LM6361 and AD848?  Was it just because
LTSpice didn't have models for them?  It can make a difference ...

Yes. LTSpice doesn't have an LM6361 or AD848. Linear lists the LT1225
as replacement for LM6361, so I thought this was safe. I deemed the AD848
to be close enough.

But as Bruce mentioned, the LT1225 is not unity gain stable, while the
LM6361 is. That might explain the oscillations.

BTW LM6361 is 550V/mv not .55V (I assume that was a mistake in your post) ...
Slew rates seem a bit different at +/- 15V supply as well.

The datasheet I have found (Natsemi Sept 1995) lists 550 V/V = 0.55 V/mV.
Which I think is a bit low, but then, the opamp is old.

		Attila Kinali

--
It is upon moral qualities that a society is ultimately founded. All
the prosperity and technological sophistication in the world is of no
use without that foundation.
-- Miss Matheson, The Diamond Age, Neil Stephenson

On Fri, 23 Feb 2018 18:48:24 -0000 "David C. Partridge" <david.partridge@perdrix.co.uk> wrote: > Looking your schematic, I see that you are using LT1225 for both U2 (U111) and U3 (U112) > > Why did you use those rather than LM6361 and AD848? Was it just because > LTSpice didn't have models for them? It can make a difference ... Yes. LTSpice doesn't have an LM6361 or AD848. Linear lists the LT1225 as replacement for LM6361, so I thought this was safe. I deemed the AD848 to be close enough. But as Bruce mentioned, the LT1225 is not unity gain stable, while the LM6361 is. That might explain the oscillations. > BTW LM6361 is 550V/mv not .55V (I assume that was a mistake in your post) ... > Slew rates seem a bit different at +/- 15V supply as well. The datasheet I have found (Natsemi Sept 1995) lists 550 V/V = 0.55 V/mV. Which I think is a bit low, but then, the opamp is old. Attila Kinali -- It is upon moral qualities that a society is ultimately founded. All the prosperity and technological sophistication in the world is of no use without that foundation. -- Miss Matheson, The Diamond Age, Neil Stephenson
BG
Bruce Griffiths
Fri, Feb 23, 2018 9:41 PM

Attila

Some insight can be gleaned by using a single pole model for an opamp configured as an integrator and deriving the expression for the output signal.
Look at the non linearity in the output.
Look at the voltage at the inverting input.
The effective impedance at the inverting input due to feedback action can be modelled as a simple serries network. What are the network parameters?
What happens to the non linearity if the positive input is driven in such a way that the inverting input is held at ground?
How would one do this?
What does the resultant signal at the inverting input look like with real opamps?
Whats the maximum GBW of the corrector amp so that the phase margin of the combination isn't reduced too much?

~30 years ago when I first analysed the limitations of opamp analog integrators I didn't have access to Spice and had to resort to Laplace transforms and analytical models of opamps.

Bruce

On 24 February 2018 at 10:23 Attila Kinali attila@kinali.ch wrote:

On Fri, 23 Feb 2018 18:48:24 -0000
"David C. Partridge" david.partridge@perdrix.co.uk wrote:

Looking your schematic, I see that you are using LT1225 for both U2 (U111) and U3 (U112)

Why did you use those rather than LM6361 and AD848?  Was it just because
LTSpice didn't have models for them?  It can make a difference ...

Yes. LTSpice doesn't have an LM6361 or AD848. Linear lists the LT1225
as replacement for LM6361, so I thought this was safe. I deemed the AD848
to be close enough.

But as Bruce mentioned, the LT1225 is not unity gain stable, while the
LM6361 is. That might explain the oscillations.

BTW LM6361 is 550V/mv not .55V (I assume that was a mistake in your post) ...
Slew rates seem a bit different at +/- 15V supply as well.

The datasheet I have found (Natsemi Sept 1995) lists 550 V/V = 0.55 V/mV.
Which I think is a bit low, but then, the opamp is old.

		Attila Kinali

--
It is upon moral qualities that a society is ultimately founded. All
the prosperity and technological sophistication in the world is of no
use without that foundation.
-- Miss Matheson, The Diamond Age, Neil Stephenson


volt-nuts mailing list -- volt-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.

Attila Some insight can be gleaned by using a single pole model for an opamp configured as an integrator and deriving the expression for the output signal. Look at the non linearity in the output. Look at the voltage at the inverting input. The effective impedance at the inverting input due to feedback action can be modelled as a simple serries network. What are the network parameters? What happens to the non linearity if the positive input is driven in such a way that the inverting input is held at ground? How would one do this? What does the resultant signal at the inverting input look like with real opamps? Whats the maximum GBW of the corrector amp so that the phase margin of the combination isn't reduced too much? ~30 years ago when I first analysed the limitations of opamp analog integrators I didn't have access to Spice and had to resort to Laplace transforms and analytical models of opamps. Bruce > On 24 February 2018 at 10:23 Attila Kinali <attila@kinali.ch> wrote: > > > On Fri, 23 Feb 2018 18:48:24 -0000 > "David C. Partridge" <david.partridge@perdrix.co.uk> wrote: > > > Looking your schematic, I see that you are using LT1225 for both U2 (U111) and U3 (U112) > > > > Why did you use those rather than LM6361 and AD848? Was it just because > > LTSpice didn't have models for them? It can make a difference ... > > Yes. LTSpice doesn't have an LM6361 or AD848. Linear lists the LT1225 > as replacement for LM6361, so I thought this was safe. I deemed the AD848 > to be close enough. > > But as Bruce mentioned, the LT1225 is not unity gain stable, while the > LM6361 is. That might explain the oscillations. > > > BTW LM6361 is 550V/mv not .55V (I assume that was a mistake in your post) ... > > Slew rates seem a bit different at +/- 15V supply as well. > > The datasheet I have found (Natsemi Sept 1995) lists 550 V/V = 0.55 V/mV. > Which I think is a bit low, but then, the opamp is old. > > Attila Kinali > -- > It is upon moral qualities that a society is ultimately founded. All > the prosperity and technological sophistication in the world is of no > use without that foundation. > -- Miss Matheson, The Diamond Age, Neil Stephenson > _______________________________________________ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there.
DC
David C. Partridge
Fri, Feb 23, 2018 10:05 PM

The datasheet I have found (Natsemi Sept 1995) lists 550 V/V = 0.55 V/mV.
Which I think is a bit low, but then, the opamp is old.

Check other datasheets, that's definitely wrong

David

> The datasheet I have found (Natsemi Sept 1995) lists 550 V/V = 0.55 V/mV. > Which I think is a bit low, but then, the opamp is old. Check other datasheets, that's definitely wrong David