time-nuts@lists.febo.com

Discussion of precise time and frequency measurement

View all threads

HP Z3801 RB fun

PS
paul swed
Wed, Dec 28, 2016 1:33 AM

OK so could not resist. The replacement of the HP 10811 with a PIEZO ocxo
went very well. So in for a penny in for a pound. Why not an RB. Having a
few around. Choice efratom FRS-c. Lot of thought went into this choice. Not
really it was there. It has a inverse frequency to EFC voltage just like
the piezo and opposite the HP 10811.
Well it connects runs and locks using the simple 1 opamp level ofset and
and phase inverter.
But locked is a funny comment. According to HPsatstat and LH. Its pll is
locked. But both scope and the LH graph shows a very slow frequency drift
that with the offset I can make positive or negative. A cycle at 10 ns
takes quite some time about an hour compared to a TBolt and another non
modified Z3801. At this level of resolution some things start to become
questionable.

I almost wonder if the modified z3801 under some level of drift assumes
everything is locked and fine. Can it be pushed beyond expectations. I
suspect it can.
Anyhow as I say a bit of GPSDO fun.
Regards
Paul
WB8TSL

OK so could not resist. The replacement of the HP 10811 with a PIEZO ocxo went very well. So in for a penny in for a pound. Why not an RB. Having a few around. Choice efratom FRS-c. Lot of thought went into this choice. Not really it was there. It has a inverse frequency to EFC voltage just like the piezo and opposite the HP 10811. Well it connects runs and locks using the simple 1 opamp level ofset and and phase inverter. But locked is a funny comment. According to HPsatstat and LH. Its pll is locked. But both scope and the LH graph shows a very slow frequency drift that with the offset I can make positive or negative. A cycle at 10 ns takes quite some time about an hour compared to a TBolt and another non modified Z3801. At this level of resolution some things start to become questionable. I almost wonder if the modified z3801 under some level of drift assumes everything is locked and fine. Can it be pushed beyond expectations. I suspect it can. Anyhow as I say a bit of GPSDO fun. Regards Paul WB8TSL
MD
Magnus Danielson
Wed, Dec 28, 2016 2:01 AM

Paul,

Have you matched up the Rb EFC sensitivity to that of the 10811.
If not, you have most probably grossly changed the loop gain, and I
can't recall that the Z3801A was designed to have different loop-gains.

Just a thought.

Cheers,
Magnus

On 12/28/2016 02:33 AM, paul swed wrote:

OK so could not resist. The replacement of the HP 10811 with a PIEZO ocxo
went very well. So in for a penny in for a pound. Why not an RB. Having a
few around. Choice efratom FRS-c. Lot of thought went into this choice. Not
really it was there. It has a inverse frequency to EFC voltage just like
the piezo and opposite the HP 10811.
Well it connects runs and locks using the simple 1 opamp level ofset and
and phase inverter.
But locked is a funny comment. According to HPsatstat and LH. Its pll is
locked. But both scope and the LH graph shows a very slow frequency drift
that with the offset I can make positive or negative. A cycle at 10 ns
takes quite some time about an hour compared to a TBolt and another non
modified Z3801. At this level of resolution some things start to become
questionable.

I almost wonder if the modified z3801 under some level of drift assumes
everything is locked and fine. Can it be pushed beyond expectations. I
suspect it can.
Anyhow as I say a bit of GPSDO fun.
Regards
Paul
WB8TSL


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Paul, Have you matched up the Rb EFC sensitivity to that of the 10811. If not, you have most probably grossly changed the loop gain, and I can't recall that the Z3801A was designed to have different loop-gains. Just a thought. Cheers, Magnus On 12/28/2016 02:33 AM, paul swed wrote: > OK so could not resist. The replacement of the HP 10811 with a PIEZO ocxo > went very well. So in for a penny in for a pound. Why not an RB. Having a > few around. Choice efratom FRS-c. Lot of thought went into this choice. Not > really it was there. It has a inverse frequency to EFC voltage just like > the piezo and opposite the HP 10811. > Well it connects runs and locks using the simple 1 opamp level ofset and > and phase inverter. > But locked is a funny comment. According to HPsatstat and LH. Its pll is > locked. But both scope and the LH graph shows a very slow frequency drift > that with the offset I can make positive or negative. A cycle at 10 ns > takes quite some time about an hour compared to a TBolt and another non > modified Z3801. At this level of resolution some things start to become > questionable. > > I almost wonder if the modified z3801 under some level of drift assumes > everything is locked and fine. Can it be pushed beyond expectations. I > suspect it can. > Anyhow as I say a bit of GPSDO fun. > Regards > Paul > WB8TSL > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
BC
Bob Camp
Wed, Dec 28, 2016 3:26 AM

Hi

The loop gain (as Magnus mentions) needs to be “correct” for the Rb. Since
the firmware was written for an OCXO with a PPM(ish) trim range, tacking it
on to a PPB(ish) Rb will require everything to scale by 1000:1. That’s 10 bits.
You may (or may not) have enough resolution in the math to handle that wide
a range of gains.

If it really locks, the next challenge is to get it to run out to ~ 4 days or more on
the loop. Again, the firmware may or may not be able to do this. Unless it does,
the Rb really will not lock as well as it could.

With the Rb correctly locked, you still have the issue that the Rb is a 2x10^-11
device (ADEV) at 1 second and the OCXO probably is 2x10^-12 (or better). For
driving a counter with a normal gate, the Rb may not show much (if any) improvement.
There is the slight chance the ADEV will get the firmware confused…..

Since the firmware source is running around somewhere, you could dig into
all of this wonderful stuff :)

Bob

On Dec 27, 2016, at 8:33 PM, paul swed paulswedb@gmail.com wrote:

OK so could not resist. The replacement of the HP 10811 with a PIEZO ocxo
went very well. So in for a penny in for a pound. Why not an RB. Having a
few around. Choice efratom FRS-c. Lot of thought went into this choice. Not
really it was there. It has a inverse frequency to EFC voltage just like
the piezo and opposite the HP 10811.
Well it connects runs and locks using the simple 1 opamp level ofset and
and phase inverter.
But locked is a funny comment. According to HPsatstat and LH. Its pll is
locked. But both scope and the LH graph shows a very slow frequency drift
that with the offset I can make positive or negative. A cycle at 10 ns
takes quite some time about an hour compared to a TBolt and another non
modified Z3801. At this level of resolution some things start to become
questionable.

I almost wonder if the modified z3801 under some level of drift assumes
everything is locked and fine. Can it be pushed beyond expectations. I
suspect it can.
Anyhow as I say a bit of GPSDO fun.
Regards
Paul
WB8TSL


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi The loop gain (as Magnus mentions) needs to be “correct” for the Rb. Since the firmware was written for an OCXO with a PPM(ish) trim range, tacking it on to a PPB(ish) Rb will require everything to scale by 1000:1. That’s 10 bits. You may (or may not) have enough resolution in the math to handle that wide a range of gains. If it really locks, the next challenge is to get it to run out to ~ 4 days or more on the loop. Again, the firmware may or may not be able to do this. Unless it does, the Rb really will not lock as well as it could. With the Rb correctly locked, you still have the issue that the Rb is a 2x10^-11 device (ADEV) at 1 second and the OCXO probably is 2x10^-12 (or better). For driving a counter with a normal gate, the Rb may not show much (if any) improvement. There is the slight chance the ADEV will get the firmware confused….. Since the firmware source is running around somewhere, you *could* dig into all of this wonderful stuff :) Bob > On Dec 27, 2016, at 8:33 PM, paul swed <paulswedb@gmail.com> wrote: > > OK so could not resist. The replacement of the HP 10811 with a PIEZO ocxo > went very well. So in for a penny in for a pound. Why not an RB. Having a > few around. Choice efratom FRS-c. Lot of thought went into this choice. Not > really it was there. It has a inverse frequency to EFC voltage just like > the piezo and opposite the HP 10811. > Well it connects runs and locks using the simple 1 opamp level ofset and > and phase inverter. > But locked is a funny comment. According to HPsatstat and LH. Its pll is > locked. But both scope and the LH graph shows a very slow frequency drift > that with the offset I can make positive or negative. A cycle at 10 ns > takes quite some time about an hour compared to a TBolt and another non > modified Z3801. At this level of resolution some things start to become > questionable. > > I almost wonder if the modified z3801 under some level of drift assumes > everything is locked and fine. Can it be pushed beyond expectations. I > suspect it can. > Anyhow as I say a bit of GPSDO fun. > Regards > Paul > WB8TSL > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
PS
paul swed
Wed, Dec 28, 2016 10:42 PM

Magnus and Bob,
Thank you for your response. As far as the original HO 10811 oscillator
goes there is some well written details on it behavior and sensitivity. It
was not the same as a standard 10811. A far smaller range as I recall. But
the data is there.
Its funny on the RB I have the EFC range from the spec sheet. It really
tunes from 0-5V. But I find it interesting that the best alignment with
other references is sub 1V. Its an old FRS C and I have others that may
present a more centered range.

I'll have to compare the detials that I have on the 10811 and FRS C. But I
actually pulled out a 100:1 antenuator I was using on the Piezo crystal
that indeed does lock.

Goal in all of this is nothing special simply curiosity.
Thanks again. When I have some real numbers to share I will.
Regards
Paul
WB8TSL

On Tue, Dec 27, 2016 at 10:26 PM, Bob Camp kb8tq@n1k.org wrote:

Hi

The loop gain (as Magnus mentions) needs to be “correct” for the Rb. Since
the firmware was written for an OCXO with a PPM(ish) trim range, tacking it
on to a PPB(ish) Rb will require everything to scale by 1000:1. That’s 10
bits.
You may (or may not) have enough resolution in the math to handle that wide
a range of gains.

If it really locks, the next challenge is to get it to run out to ~ 4 days
or more on
the loop. Again, the firmware may or may not be able to do this. Unless it
does,
the Rb really will not lock as well as it could.

With the Rb correctly locked, you still have the issue that the Rb is a
2x10^-11
device (ADEV) at 1 second and the OCXO probably is 2x10^-12 (or better).
For
driving a counter with a normal gate, the Rb may not show much (if any)
improvement.
There is the slight chance the ADEV will get the firmware confused…..

Since the firmware source is running around somewhere, you could dig into
all of this wonderful stuff :)

Bob

On Dec 27, 2016, at 8:33 PM, paul swed paulswedb@gmail.com wrote:

OK so could not resist. The replacement of the HP 10811 with a PIEZO ocxo
went very well. So in for a penny in for a pound. Why not an RB. Having a
few around. Choice efratom FRS-c. Lot of thought went into this choice.

Not

really it was there. It has a inverse frequency to EFC voltage just like
the piezo and opposite the HP 10811.
Well it connects runs and locks using the simple 1 opamp level ofset and
and phase inverter.
But locked is a funny comment. According to HPsatstat and LH. Its pll is
locked. But both scope and the LH graph shows a very slow frequency drift
that with the offset I can make positive or negative. A cycle at 10 ns
takes quite some time about an hour compared to a TBolt and another non
modified Z3801. At this level of resolution some things start to become
questionable.

I almost wonder if the modified z3801 under some level of drift assumes
everything is locked and fine. Can it be pushed beyond expectations. I
suspect it can.
Anyhow as I say a bit of GPSDO fun.
Regards
Paul
WB8TSL


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

Magnus and Bob, Thank you for your response. As far as the original HO 10811 oscillator goes there is some well written details on it behavior and sensitivity. It was not the same as a standard 10811. A far smaller range as I recall. But the data is there. Its funny on the RB I have the EFC range from the spec sheet. It really tunes from 0-5V. But I find it interesting that the best alignment with other references is sub 1V. Its an old FRS C and I have others that may present a more centered range. I'll have to compare the detials that I have on the 10811 and FRS C. But I actually pulled out a 100:1 antenuator I was using on the Piezo crystal that indeed does lock. Goal in all of this is nothing special simply curiosity. Thanks again. When I have some real numbers to share I will. Regards Paul WB8TSL On Tue, Dec 27, 2016 at 10:26 PM, Bob Camp <kb8tq@n1k.org> wrote: > Hi > > The loop gain (as Magnus mentions) needs to be “correct” for the Rb. Since > the firmware was written for an OCXO with a PPM(ish) trim range, tacking it > on to a PPB(ish) Rb will require everything to scale by 1000:1. That’s 10 > bits. > You may (or may not) have enough resolution in the math to handle that wide > a range of gains. > > If it really locks, the next challenge is to get it to run out to ~ 4 days > or more on > the loop. Again, the firmware may or may not be able to do this. Unless it > does, > the Rb really will not lock as well as it could. > > With the Rb correctly locked, you still have the issue that the Rb is a > 2x10^-11 > device (ADEV) at 1 second and the OCXO probably is 2x10^-12 (or better). > For > driving a counter with a normal gate, the Rb may not show much (if any) > improvement. > There is the slight chance the ADEV will get the firmware confused….. > > Since the firmware source is running around somewhere, you *could* dig into > all of this wonderful stuff :) > > Bob > > > > On Dec 27, 2016, at 8:33 PM, paul swed <paulswedb@gmail.com> wrote: > > > > OK so could not resist. The replacement of the HP 10811 with a PIEZO ocxo > > went very well. So in for a penny in for a pound. Why not an RB. Having a > > few around. Choice efratom FRS-c. Lot of thought went into this choice. > Not > > really it was there. It has a inverse frequency to EFC voltage just like > > the piezo and opposite the HP 10811. > > Well it connects runs and locks using the simple 1 opamp level ofset and > > and phase inverter. > > But locked is a funny comment. According to HPsatstat and LH. Its pll is > > locked. But both scope and the LH graph shows a very slow frequency drift > > that with the offset I can make positive or negative. A cycle at 10 ns > > takes quite some time about an hour compared to a TBolt and another non > > modified Z3801. At this level of resolution some things start to become > > questionable. > > > > I almost wonder if the modified z3801 under some level of drift assumes > > everything is locked and fine. Can it be pushed beyond expectations. I > > suspect it can. > > Anyhow as I say a bit of GPSDO fun. > > Regards > > Paul > > WB8TSL > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > and follow the instructions there. >
MD
Magnus Danielson
Wed, Dec 28, 2016 11:22 PM

Paul,

We just want to help you with that curiosity, it sounds like a fun
little project. :)

The FRS-C [1] has a range of +/- 1E-9 over 0-5V, so EFC sensitivity is
4E-10/V.

The Z3801A has a range of 5.20E-10 [2] over -5V to +5V [3], so EFC
sensitivity is 5.2E-11/V.

So, it looks like you need to have a gain of 8 and raise the EFC 2.5 V.

Cheers,
Magnus

[1] http://www.to-way.com/tf/frs.pdf

[2] http://www.leapsecond.com/pages/z3801a-efc/

[3] http://www.realhamradio.com/joe-geller.htm

On 12/28/2016 11:42 PM, paul swed wrote:

Magnus and Bob,
Thank you for your response. As far as the original HO 10811 oscillator
goes there is some well written details on it behavior and sensitivity. It
was not the same as a standard 10811. A far smaller range as I recall. But
the data is there.
Its funny on the RB I have the EFC range from the spec sheet. It really
tunes from 0-5V. But I find it interesting that the best alignment with
other references is sub 1V. Its an old FRS C and I have others that may
present a more centered range.

I'll have to compare the detials that I have on the 10811 and FRS C. But I
actually pulled out a 100:1 antenuator I was using on the Piezo crystal
that indeed does lock.

Goal in all of this is nothing special simply curiosity.
Thanks again. When I have some real numbers to share I will.
Regards
Paul
WB8TSL

On Tue, Dec 27, 2016 at 10:26 PM, Bob Camp kb8tq@n1k.org wrote:

Hi

The loop gain (as Magnus mentions) needs to be “correct” for the Rb. Since
the firmware was written for an OCXO with a PPM(ish) trim range, tacking it
on to a PPB(ish) Rb will require everything to scale by 1000:1. That’s 10
bits.
You may (or may not) have enough resolution in the math to handle that wide
a range of gains.

If it really locks, the next challenge is to get it to run out to ~ 4 days
or more on
the loop. Again, the firmware may or may not be able to do this. Unless it
does,
the Rb really will not lock as well as it could.

With the Rb correctly locked, you still have the issue that the Rb is a
2x10^-11
device (ADEV) at 1 second and the OCXO probably is 2x10^-12 (or better).
For
driving a counter with a normal gate, the Rb may not show much (if any)
improvement.
There is the slight chance the ADEV will get the firmware confused…..

Since the firmware source is running around somewhere, you could dig into
all of this wonderful stuff :)

Bob

On Dec 27, 2016, at 8:33 PM, paul swed paulswedb@gmail.com wrote:

OK so could not resist. The replacement of the HP 10811 with a PIEZO ocxo
went very well. So in for a penny in for a pound. Why not an RB. Having a
few around. Choice efratom FRS-c. Lot of thought went into this choice.

Not

really it was there. It has a inverse frequency to EFC voltage just like
the piezo and opposite the HP 10811.
Well it connects runs and locks using the simple 1 opamp level ofset and
and phase inverter.
But locked is a funny comment. According to HPsatstat and LH. Its pll is
locked. But both scope and the LH graph shows a very slow frequency drift
that with the offset I can make positive or negative. A cycle at 10 ns
takes quite some time about an hour compared to a TBolt and another non
modified Z3801. At this level of resolution some things start to become
questionable.

I almost wonder if the modified z3801 under some level of drift assumes
everything is locked and fine. Can it be pushed beyond expectations. I
suspect it can.
Anyhow as I say a bit of GPSDO fun.
Regards
Paul
WB8TSL


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Paul, We just want to help you with that curiosity, it sounds like a fun little project. :) The FRS-C [1] has a range of +/- 1E-9 over 0-5V, so EFC sensitivity is 4E-10/V. The Z3801A has a range of 5.20E-10 [2] over -5V to +5V [3], so EFC sensitivity is 5.2E-11/V. So, it looks like you need to have a gain of 8 and raise the EFC 2.5 V. Cheers, Magnus [1] http://www.to-way.com/tf/frs.pdf [2] http://www.leapsecond.com/pages/z3801a-efc/ [3] http://www.realhamradio.com/joe-geller.htm On 12/28/2016 11:42 PM, paul swed wrote: > Magnus and Bob, > Thank you for your response. As far as the original HO 10811 oscillator > goes there is some well written details on it behavior and sensitivity. It > was not the same as a standard 10811. A far smaller range as I recall. But > the data is there. > Its funny on the RB I have the EFC range from the spec sheet. It really > tunes from 0-5V. But I find it interesting that the best alignment with > other references is sub 1V. Its an old FRS C and I have others that may > present a more centered range. > > I'll have to compare the detials that I have on the 10811 and FRS C. But I > actually pulled out a 100:1 antenuator I was using on the Piezo crystal > that indeed does lock. > > Goal in all of this is nothing special simply curiosity. > Thanks again. When I have some real numbers to share I will. > Regards > Paul > WB8TSL > > On Tue, Dec 27, 2016 at 10:26 PM, Bob Camp <kb8tq@n1k.org> wrote: > >> Hi >> >> The loop gain (as Magnus mentions) needs to be “correct” for the Rb. Since >> the firmware was written for an OCXO with a PPM(ish) trim range, tacking it >> on to a PPB(ish) Rb will require everything to scale by 1000:1. That’s 10 >> bits. >> You may (or may not) have enough resolution in the math to handle that wide >> a range of gains. >> >> If it really locks, the next challenge is to get it to run out to ~ 4 days >> or more on >> the loop. Again, the firmware may or may not be able to do this. Unless it >> does, >> the Rb really will not lock as well as it could. >> >> With the Rb correctly locked, you still have the issue that the Rb is a >> 2x10^-11 >> device (ADEV) at 1 second and the OCXO probably is 2x10^-12 (or better). >> For >> driving a counter with a normal gate, the Rb may not show much (if any) >> improvement. >> There is the slight chance the ADEV will get the firmware confused….. >> >> Since the firmware source is running around somewhere, you *could* dig into >> all of this wonderful stuff :) >> >> Bob >> >> >>> On Dec 27, 2016, at 8:33 PM, paul swed <paulswedb@gmail.com> wrote: >>> >>> OK so could not resist. The replacement of the HP 10811 with a PIEZO ocxo >>> went very well. So in for a penny in for a pound. Why not an RB. Having a >>> few around. Choice efratom FRS-c. Lot of thought went into this choice. >> Not >>> really it was there. It has a inverse frequency to EFC voltage just like >>> the piezo and opposite the HP 10811. >>> Well it connects runs and locks using the simple 1 opamp level ofset and >>> and phase inverter. >>> But locked is a funny comment. According to HPsatstat and LH. Its pll is >>> locked. But both scope and the LH graph shows a very slow frequency drift >>> that with the offset I can make positive or negative. A cycle at 10 ns >>> takes quite some time about an hour compared to a TBolt and another non >>> modified Z3801. At this level of resolution some things start to become >>> questionable. >>> >>> I almost wonder if the modified z3801 under some level of drift assumes >>> everything is locked and fine. Can it be pushed beyond expectations. I >>> suspect it can. >>> Anyhow as I say a bit of GPSDO fun. >>> Regards >>> Paul >>> WB8TSL >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >> mailman/listinfo/time-nuts >>> and follow the instructions there. >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/ >> mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
BC
Bob Camp
Wed, Dec 28, 2016 11:32 PM

Hi

Ummm ….. errrrr ….

The 10811 that is used in the Z3801 has a larger electrical tuning range than the standard
10811 not a smaller one. It’s tuning range and the TBolt OCXO’s tuning range are
very similar.

Bob

On Dec 28, 2016, at 6:22 PM, Magnus Danielson magnus@rubidium.dyndns.org wrote:

Paul,

We just want to help you with that curiosity, it sounds like a fun little project. :)

The FRS-C [1] has a range of +/- 1E-9 over 0-5V, so EFC sensitivity is 4E-10/V.

The Z3801A has a range of 5.20E-10 [2] over -5V to +5V [3], so EFC sensitivity is 5.2E-11/V.

So, it looks like you need to have a gain of 8 and raise the EFC 2.5 V.

Cheers,
Magnus

[1] http://www.to-way.com/tf/frs.pdf

[2] http://www.leapsecond.com/pages/z3801a-efc/

[3] http://www.realhamradio.com/joe-geller.htm

On 12/28/2016 11:42 PM, paul swed wrote:

Magnus and Bob,
Thank you for your response. As far as the original HO 10811 oscillator
goes there is some well written details on it behavior and sensitivity. It
was not the same as a standard 10811. A far smaller range as I recall. But
the data is there.
Its funny on the RB I have the EFC range from the spec sheet. It really
tunes from 0-5V. But I find it interesting that the best alignment with
other references is sub 1V. Its an old FRS C and I have others that may
present a more centered range.

I'll have to compare the detials that I have on the 10811 and FRS C. But I
actually pulled out a 100:1 antenuator I was using on the Piezo crystal
that indeed does lock.

Goal in all of this is nothing special simply curiosity.
Thanks again. When I have some real numbers to share I will.
Regards
Paul
WB8TSL

On Tue, Dec 27, 2016 at 10:26 PM, Bob Camp kb8tq@n1k.org wrote:

Hi

The loop gain (as Magnus mentions) needs to be “correct” for the Rb. Since
the firmware was written for an OCXO with a PPM(ish) trim range, tacking it
on to a PPB(ish) Rb will require everything to scale by 1000:1. That’s 10
bits.
You may (or may not) have enough resolution in the math to handle that wide
a range of gains.

If it really locks, the next challenge is to get it to run out to ~ 4 days
or more on
the loop. Again, the firmware may or may not be able to do this. Unless it
does,
the Rb really will not lock as well as it could.

With the Rb correctly locked, you still have the issue that the Rb is a
2x10^-11
device (ADEV) at 1 second and the OCXO probably is 2x10^-12 (or better).
For
driving a counter with a normal gate, the Rb may not show much (if any)
improvement.
There is the slight chance the ADEV will get the firmware confused…..

Since the firmware source is running around somewhere, you could dig into
all of this wonderful stuff :)

Bob

On Dec 27, 2016, at 8:33 PM, paul swed paulswedb@gmail.com wrote:

OK so could not resist. The replacement of the HP 10811 with a PIEZO ocxo
went very well. So in for a penny in for a pound. Why not an RB. Having a
few around. Choice efratom FRS-c. Lot of thought went into this choice.

Not

really it was there. It has a inverse frequency to EFC voltage just like
the piezo and opposite the HP 10811.
Well it connects runs and locks using the simple 1 opamp level ofset and
and phase inverter.
But locked is a funny comment. According to HPsatstat and LH. Its pll is
locked. But both scope and the LH graph shows a very slow frequency drift
that with the offset I can make positive or negative. A cycle at 10 ns
takes quite some time about an hour compared to a TBolt and another non
modified Z3801. At this level of resolution some things start to become
questionable.

I almost wonder if the modified z3801 under some level of drift assumes
everything is locked and fine. Can it be pushed beyond expectations. I
suspect it can.
Anyhow as I say a bit of GPSDO fun.
Regards
Paul
WB8TSL


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi Ummm ….. errrrr …. The 10811 that is used in the Z3801 has a *larger* electrical tuning range than the standard 10811 not a smaller one. It’s tuning range and the TBolt OCXO’s tuning range are very similar. Bob > On Dec 28, 2016, at 6:22 PM, Magnus Danielson <magnus@rubidium.dyndns.org> wrote: > > Paul, > > We just want to help you with that curiosity, it sounds like a fun little project. :) > > The FRS-C [1] has a range of +/- 1E-9 over 0-5V, so EFC sensitivity is 4E-10/V. > > The Z3801A has a range of 5.20E-10 [2] over -5V to +5V [3], so EFC sensitivity is 5.2E-11/V. > > So, it looks like you need to have a gain of 8 and raise the EFC 2.5 V. > > Cheers, > Magnus > > [1] http://www.to-way.com/tf/frs.pdf > > [2] http://www.leapsecond.com/pages/z3801a-efc/ > > [3] http://www.realhamradio.com/joe-geller.htm > > On 12/28/2016 11:42 PM, paul swed wrote: >> Magnus and Bob, >> Thank you for your response. As far as the original HO 10811 oscillator >> goes there is some well written details on it behavior and sensitivity. It >> was not the same as a standard 10811. A far smaller range as I recall. But >> the data is there. >> Its funny on the RB I have the EFC range from the spec sheet. It really >> tunes from 0-5V. But I find it interesting that the best alignment with >> other references is sub 1V. Its an old FRS C and I have others that may >> present a more centered range. >> >> I'll have to compare the detials that I have on the 10811 and FRS C. But I >> actually pulled out a 100:1 antenuator I was using on the Piezo crystal >> that indeed does lock. >> >> Goal in all of this is nothing special simply curiosity. >> Thanks again. When I have some real numbers to share I will. >> Regards >> Paul >> WB8TSL >> >> On Tue, Dec 27, 2016 at 10:26 PM, Bob Camp <kb8tq@n1k.org> wrote: >> >>> Hi >>> >>> The loop gain (as Magnus mentions) needs to be “correct” for the Rb. Since >>> the firmware was written for an OCXO with a PPM(ish) trim range, tacking it >>> on to a PPB(ish) Rb will require everything to scale by 1000:1. That’s 10 >>> bits. >>> You may (or may not) have enough resolution in the math to handle that wide >>> a range of gains. >>> >>> If it really locks, the next challenge is to get it to run out to ~ 4 days >>> or more on >>> the loop. Again, the firmware may or may not be able to do this. Unless it >>> does, >>> the Rb really will not lock as well as it could. >>> >>> With the Rb correctly locked, you still have the issue that the Rb is a >>> 2x10^-11 >>> device (ADEV) at 1 second and the OCXO probably is 2x10^-12 (or better). >>> For >>> driving a counter with a normal gate, the Rb may not show much (if any) >>> improvement. >>> There is the slight chance the ADEV will get the firmware confused….. >>> >>> Since the firmware source is running around somewhere, you *could* dig into >>> all of this wonderful stuff :) >>> >>> Bob >>> >>> >>>> On Dec 27, 2016, at 8:33 PM, paul swed <paulswedb@gmail.com> wrote: >>>> >>>> OK so could not resist. The replacement of the HP 10811 with a PIEZO ocxo >>>> went very well. So in for a penny in for a pound. Why not an RB. Having a >>>> few around. Choice efratom FRS-c. Lot of thought went into this choice. >>> Not >>>> really it was there. It has a inverse frequency to EFC voltage just like >>>> the piezo and opposite the HP 10811. >>>> Well it connects runs and locks using the simple 1 opamp level ofset and >>>> and phase inverter. >>>> But locked is a funny comment. According to HPsatstat and LH. Its pll is >>>> locked. But both scope and the LH graph shows a very slow frequency drift >>>> that with the offset I can make positive or negative. A cycle at 10 ns >>>> takes quite some time about an hour compared to a TBolt and another non >>>> modified Z3801. At this level of resolution some things start to become >>>> questionable. >>>> >>>> I almost wonder if the modified z3801 under some level of drift assumes >>>> everything is locked and fine. Can it be pushed beyond expectations. I >>>> suspect it can. >>>> Anyhow as I say a bit of GPSDO fun. >>>> Regards >>>> Paul >>>> WB8TSL >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>> mailman/listinfo/time-nuts >>>> and follow the instructions there. >>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>> mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. >> > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there.
PS
paul swed
Wed, Dec 28, 2016 11:48 PM

​Magnus and Bob.
I appreciate the comments and the suggestions are easily implemented.
Technical term "Piece-O-cake".
One other note the the Z3801 uses a more negative voltage to raise
frequency while the Piezo and FRS use a positive voltage.
Absolutely agree on the offset​. Been trying to do that. However the FRS C
I am using wants seems to require +.6V to be on frequency. So because of
its age there may be a real issue. I have 3 others and may grab another to
see how it behaves. Also have some EGGs and FEI 5680s. Though those as I
recall are digital control.
But back to the basics first.
Regards
Paul.
WB8TSL

On Wed, Dec 28, 2016 at 6:32 PM, Bob Camp kb8tq@n1k.org wrote:

Hi

Ummm ….. errrrr ….

The 10811 that is used in the Z3801 has a larger electrical tuning range
than the standard
10811 not a smaller one. It’s tuning range and the TBolt OCXO’s tuning
range are
very similar.

Bob

On Dec 28, 2016, at 6:22 PM, Magnus Danielson <

Paul,

We just want to help you with that curiosity, it sounds like a fun

little project. :)

The FRS-C [1] has a range of +/- 1E-9 over 0-5V, so EFC sensitivity is

4E-10/V.

The Z3801A has a range of 5.20E-10 [2] over -5V to +5V [3], so EFC

sensitivity is 5.2E-11/V.

So, it looks like you need to have a gain of 8 and raise the EFC 2.5 V.

Cheers,
Magnus

[1] http://www.to-way.com/tf/frs.pdf

[2] http://www.leapsecond.com/pages/z3801a-efc/

[3] http://www.realhamradio.com/joe-geller.htm

On 12/28/2016 11:42 PM, paul swed wrote:

Magnus and Bob,
Thank you for your response. As far as the original HO 10811 oscillator
goes there is some well written details on it behavior and sensitivity.

It

was not the same as a standard 10811. A far smaller range as I recall.

But

the data is there.
Its funny on the RB I have the EFC range from the spec sheet. It really
tunes from 0-5V. But I find it interesting that the best alignment with
other references is sub 1V. Its an old FRS C and I have others that may
present a more centered range.

I'll have to compare the detials that I have on the 10811 and FRS C.

But I

actually pulled out a 100:1 antenuator I was using on the Piezo crystal
that indeed does lock.

Goal in all of this is nothing special simply curiosity.
Thanks again. When I have some real numbers to share I will.
Regards
Paul
WB8TSL

On Tue, Dec 27, 2016 at 10:26 PM, Bob Camp kb8tq@n1k.org wrote:

Hi

The loop gain (as Magnus mentions) needs to be “correct” for the Rb.

Since

the firmware was written for an OCXO with a PPM(ish) trim range,

tacking it

on to a PPB(ish) Rb will require everything to scale by 1000:1. That’s

10

bits.
You may (or may not) have enough resolution in the math to handle that

wide

a range of gains.

If it really locks, the next challenge is to get it to run out to ~ 4

days

or more on
the loop. Again, the firmware may or may not be able to do this.

Unless it

does,
the Rb really will not lock as well as it could.

With the Rb correctly locked, you still have the issue that the Rb is a
2x10^-11
device (ADEV) at 1 second and the OCXO probably is 2x10^-12 (or

better).

For
driving a counter with a normal gate, the Rb may not show much (if any)
improvement.
There is the slight chance the ADEV will get the firmware confused…..

Since the firmware source is running around somewhere, you could dig

into

all of this wonderful stuff :)

Bob

On Dec 27, 2016, at 8:33 PM, paul swed paulswedb@gmail.com wrote:

OK so could not resist. The replacement of the HP 10811 with a PIEZO

ocxo

went very well. So in for a penny in for a pound. Why not an RB.

Having a

few around. Choice efratom FRS-c. Lot of thought went into this

choice.

Not

really it was there. It has a inverse frequency to EFC voltage just

like

the piezo and opposite the HP 10811.
Well it connects runs and locks using the simple 1 opamp level ofset

and

and phase inverter.
But locked is a funny comment. According to HPsatstat and LH. Its pll

is

locked. But both scope and the LH graph shows a very slow frequency

drift

that with the offset I can make positive or negative. A cycle at 10 ns
takes quite some time about an hour compared to a TBolt and another

non

modified Z3801. At this level of resolution some things start to

become

questionable.

I almost wonder if the modified z3801 under some level of drift

assumes

everything is locked and fine. Can it be pushed beyond expectations. I
suspect it can.
Anyhow as I say a bit of GPSDO fun.
Regards
Paul
WB8TSL


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.

​Magnus and Bob. I appreciate the comments and the suggestions are easily implemented. Technical term "Piece-O-cake". One other note the the Z3801 uses a more negative voltage to raise frequency while the Piezo and FRS use a positive voltage. Absolutely agree on the offset​. Been trying to do that. However the FRS C I am using wants seems to require +.6V to be on frequency. So because of its age there may be a real issue. I have 3 others and may grab another to see how it behaves. Also have some EGGs and FEI 5680s. Though those as I recall are digital control. But back to the basics first. Regards Paul. WB8TSL On Wed, Dec 28, 2016 at 6:32 PM, Bob Camp <kb8tq@n1k.org> wrote: > Hi > > Ummm ….. errrrr …. > > The 10811 that is used in the Z3801 has a *larger* electrical tuning range > than the standard > 10811 not a smaller one. It’s tuning range and the TBolt OCXO’s tuning > range are > very similar. > > Bob > > > On Dec 28, 2016, at 6:22 PM, Magnus Danielson < > magnus@rubidium.dyndns.org> wrote: > > > > Paul, > > > > We just want to help you with that curiosity, it sounds like a fun > little project. :) > > > > The FRS-C [1] has a range of +/- 1E-9 over 0-5V, so EFC sensitivity is > 4E-10/V. > > > > The Z3801A has a range of 5.20E-10 [2] over -5V to +5V [3], so EFC > sensitivity is 5.2E-11/V. > > > > So, it looks like you need to have a gain of 8 and raise the EFC 2.5 V. > > > > Cheers, > > Magnus > > > > [1] http://www.to-way.com/tf/frs.pdf > > > > [2] http://www.leapsecond.com/pages/z3801a-efc/ > > > > [3] http://www.realhamradio.com/joe-geller.htm > > > > On 12/28/2016 11:42 PM, paul swed wrote: > >> Magnus and Bob, > >> Thank you for your response. As far as the original HO 10811 oscillator > >> goes there is some well written details on it behavior and sensitivity. > It > >> was not the same as a standard 10811. A far smaller range as I recall. > But > >> the data is there. > >> Its funny on the RB I have the EFC range from the spec sheet. It really > >> tunes from 0-5V. But I find it interesting that the best alignment with > >> other references is sub 1V. Its an old FRS C and I have others that may > >> present a more centered range. > >> > >> I'll have to compare the detials that I have on the 10811 and FRS C. > But I > >> actually pulled out a 100:1 antenuator I was using on the Piezo crystal > >> that indeed does lock. > >> > >> Goal in all of this is nothing special simply curiosity. > >> Thanks again. When I have some real numbers to share I will. > >> Regards > >> Paul > >> WB8TSL > >> > >> On Tue, Dec 27, 2016 at 10:26 PM, Bob Camp <kb8tq@n1k.org> wrote: > >> > >>> Hi > >>> > >>> The loop gain (as Magnus mentions) needs to be “correct” for the Rb. > Since > >>> the firmware was written for an OCXO with a PPM(ish) trim range, > tacking it > >>> on to a PPB(ish) Rb will require everything to scale by 1000:1. That’s > 10 > >>> bits. > >>> You may (or may not) have enough resolution in the math to handle that > wide > >>> a range of gains. > >>> > >>> If it really locks, the next challenge is to get it to run out to ~ 4 > days > >>> or more on > >>> the loop. Again, the firmware may or may not be able to do this. > Unless it > >>> does, > >>> the Rb really will not lock as well as it could. > >>> > >>> With the Rb correctly locked, you still have the issue that the Rb is a > >>> 2x10^-11 > >>> device (ADEV) at 1 second and the OCXO probably is 2x10^-12 (or > better). > >>> For > >>> driving a counter with a normal gate, the Rb may not show much (if any) > >>> improvement. > >>> There is the slight chance the ADEV will get the firmware confused….. > >>> > >>> Since the firmware source is running around somewhere, you *could* dig > into > >>> all of this wonderful stuff :) > >>> > >>> Bob > >>> > >>> > >>>> On Dec 27, 2016, at 8:33 PM, paul swed <paulswedb@gmail.com> wrote: > >>>> > >>>> OK so could not resist. The replacement of the HP 10811 with a PIEZO > ocxo > >>>> went very well. So in for a penny in for a pound. Why not an RB. > Having a > >>>> few around. Choice efratom FRS-c. Lot of thought went into this > choice. > >>> Not > >>>> really it was there. It has a inverse frequency to EFC voltage just > like > >>>> the piezo and opposite the HP 10811. > >>>> Well it connects runs and locks using the simple 1 opamp level ofset > and > >>>> and phase inverter. > >>>> But locked is a funny comment. According to HPsatstat and LH. Its pll > is > >>>> locked. But both scope and the LH graph shows a very slow frequency > drift > >>>> that with the offset I can make positive or negative. A cycle at 10 ns > >>>> takes quite some time about an hour compared to a TBolt and another > non > >>>> modified Z3801. At this level of resolution some things start to > become > >>>> questionable. > >>>> > >>>> I almost wonder if the modified z3801 under some level of drift > assumes > >>>> everything is locked and fine. Can it be pushed beyond expectations. I > >>>> suspect it can. > >>>> Anyhow as I say a bit of GPSDO fun. > >>>> Regards > >>>> Paul > >>>> WB8TSL > >>>> _______________________________________________ > >>>> time-nuts mailing list -- time-nuts@febo.com > >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ > >>> mailman/listinfo/time-nuts > >>>> and follow the instructions there. > >>> > >>> _______________________________________________ > >>> time-nuts mailing list -- time-nuts@febo.com > >>> To unsubscribe, go to https://www.febo.com/cgi-bin/ > >>> mailman/listinfo/time-nuts > >>> and follow the instructions there. > >>> > >> _______________________________________________ > >> time-nuts mailing list -- time-nuts@febo.com > >> To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > >> and follow the instructions there. > >> > > _______________________________________________ > > time-nuts mailing list -- time-nuts@febo.com > > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > and follow the instructions there. >
MD
Magnus Danielson
Wed, Dec 28, 2016 11:51 PM

Hi,

Well, that was the data I was able to find (and I referred to TvBs page
where I found it). If you have more accurate data, please share so we
can get the numbers right.

The Z3801A will do a least square estimate of frequency error and then
translate that into an initial DAC setting assuming it knows the EFC
sensitivity, and the loop will work out the rest from there (hopefully).
Getting EFC in the right neighborhood is probably wise for this reason.

Cheers,
Magnus

On 12/29/2016 12:32 AM, Bob Camp wrote:

Hi

Ummm ….. errrrr ….

The 10811 that is used in the Z3801 has a larger electrical tuning range than the standard
10811 not a smaller one. It’s tuning range and the TBolt OCXO’s tuning range are
very similar.

Bob

On Dec 28, 2016, at 6:22 PM, Magnus Danielson magnus@rubidium.dyndns.org wrote:

Paul,

We just want to help you with that curiosity, it sounds like a fun little project. :)

The FRS-C [1] has a range of +/- 1E-9 over 0-5V, so EFC sensitivity is 4E-10/V.

The Z3801A has a range of 5.20E-10 [2] over -5V to +5V [3], so EFC sensitivity is 5.2E-11/V.

So, it looks like you need to have a gain of 8 and raise the EFC 2.5 V.

Cheers,
Magnus

[1] http://www.to-way.com/tf/frs.pdf

[2] http://www.leapsecond.com/pages/z3801a-efc/

[3] http://www.realhamradio.com/joe-geller.htm

On 12/28/2016 11:42 PM, paul swed wrote:

Magnus and Bob,
Thank you for your response. As far as the original HO 10811 oscillator
goes there is some well written details on it behavior and sensitivity. It
was not the same as a standard 10811. A far smaller range as I recall. But
the data is there.
Its funny on the RB I have the EFC range from the spec sheet. It really
tunes from 0-5V. But I find it interesting that the best alignment with
other references is sub 1V. Its an old FRS C and I have others that may
present a more centered range.

I'll have to compare the detials that I have on the 10811 and FRS C. But I
actually pulled out a 100:1 antenuator I was using on the Piezo crystal
that indeed does lock.

Goal in all of this is nothing special simply curiosity.
Thanks again. When I have some real numbers to share I will.
Regards
Paul
WB8TSL

On Tue, Dec 27, 2016 at 10:26 PM, Bob Camp kb8tq@n1k.org wrote:

Hi

The loop gain (as Magnus mentions) needs to be “correct” for the Rb. Since
the firmware was written for an OCXO with a PPM(ish) trim range, tacking it
on to a PPB(ish) Rb will require everything to scale by 1000:1. That’s 10
bits.
You may (or may not) have enough resolution in the math to handle that wide
a range of gains.

If it really locks, the next challenge is to get it to run out to ~ 4 days
or more on
the loop. Again, the firmware may or may not be able to do this. Unless it
does,
the Rb really will not lock as well as it could.

With the Rb correctly locked, you still have the issue that the Rb is a
2x10^-11
device (ADEV) at 1 second and the OCXO probably is 2x10^-12 (or better).
For
driving a counter with a normal gate, the Rb may not show much (if any)
improvement.
There is the slight chance the ADEV will get the firmware confused…..

Since the firmware source is running around somewhere, you could dig into
all of this wonderful stuff :)

Bob

On Dec 27, 2016, at 8:33 PM, paul swed paulswedb@gmail.com wrote:

OK so could not resist. The replacement of the HP 10811 with a PIEZO ocxo
went very well. So in for a penny in for a pound. Why not an RB. Having a
few around. Choice efratom FRS-c. Lot of thought went into this choice.

Not

really it was there. It has a inverse frequency to EFC voltage just like
the piezo and opposite the HP 10811.
Well it connects runs and locks using the simple 1 opamp level ofset and
and phase inverter.
But locked is a funny comment. According to HPsatstat and LH. Its pll is
locked. But both scope and the LH graph shows a very slow frequency drift
that with the offset I can make positive or negative. A cycle at 10 ns
takes quite some time about an hour compared to a TBolt and another non
modified Z3801. At this level of resolution some things start to become
questionable.

I almost wonder if the modified z3801 under some level of drift assumes
everything is locked and fine. Can it be pushed beyond expectations. I
suspect it can.
Anyhow as I say a bit of GPSDO fun.
Regards
Paul
WB8TSL


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Hi, Well, that was the data I was able to find (and I referred to TvBs page where I found it). If you have more accurate data, please share so we can get the numbers right. The Z3801A will do a least square estimate of frequency error and then translate that into an initial DAC setting assuming it knows the EFC sensitivity, and the loop will work out the rest from there (hopefully). Getting EFC in the right neighborhood is probably wise for this reason. Cheers, Magnus On 12/29/2016 12:32 AM, Bob Camp wrote: > Hi > > Ummm ….. errrrr …. > > The 10811 that is used in the Z3801 has a *larger* electrical tuning range than the standard > 10811 not a smaller one. It’s tuning range and the TBolt OCXO’s tuning range are > very similar. > > Bob > >> On Dec 28, 2016, at 6:22 PM, Magnus Danielson <magnus@rubidium.dyndns.org> wrote: >> >> Paul, >> >> We just want to help you with that curiosity, it sounds like a fun little project. :) >> >> The FRS-C [1] has a range of +/- 1E-9 over 0-5V, so EFC sensitivity is 4E-10/V. >> >> The Z3801A has a range of 5.20E-10 [2] over -5V to +5V [3], so EFC sensitivity is 5.2E-11/V. >> >> So, it looks like you need to have a gain of 8 and raise the EFC 2.5 V. >> >> Cheers, >> Magnus >> >> [1] http://www.to-way.com/tf/frs.pdf >> >> [2] http://www.leapsecond.com/pages/z3801a-efc/ >> >> [3] http://www.realhamradio.com/joe-geller.htm >> >> On 12/28/2016 11:42 PM, paul swed wrote: >>> Magnus and Bob, >>> Thank you for your response. As far as the original HO 10811 oscillator >>> goes there is some well written details on it behavior and sensitivity. It >>> was not the same as a standard 10811. A far smaller range as I recall. But >>> the data is there. >>> Its funny on the RB I have the EFC range from the spec sheet. It really >>> tunes from 0-5V. But I find it interesting that the best alignment with >>> other references is sub 1V. Its an old FRS C and I have others that may >>> present a more centered range. >>> >>> I'll have to compare the detials that I have on the 10811 and FRS C. But I >>> actually pulled out a 100:1 antenuator I was using on the Piezo crystal >>> that indeed does lock. >>> >>> Goal in all of this is nothing special simply curiosity. >>> Thanks again. When I have some real numbers to share I will. >>> Regards >>> Paul >>> WB8TSL >>> >>> On Tue, Dec 27, 2016 at 10:26 PM, Bob Camp <kb8tq@n1k.org> wrote: >>> >>>> Hi >>>> >>>> The loop gain (as Magnus mentions) needs to be “correct” for the Rb. Since >>>> the firmware was written for an OCXO with a PPM(ish) trim range, tacking it >>>> on to a PPB(ish) Rb will require everything to scale by 1000:1. That’s 10 >>>> bits. >>>> You may (or may not) have enough resolution in the math to handle that wide >>>> a range of gains. >>>> >>>> If it really locks, the next challenge is to get it to run out to ~ 4 days >>>> or more on >>>> the loop. Again, the firmware may or may not be able to do this. Unless it >>>> does, >>>> the Rb really will not lock as well as it could. >>>> >>>> With the Rb correctly locked, you still have the issue that the Rb is a >>>> 2x10^-11 >>>> device (ADEV) at 1 second and the OCXO probably is 2x10^-12 (or better). >>>> For >>>> driving a counter with a normal gate, the Rb may not show much (if any) >>>> improvement. >>>> There is the slight chance the ADEV will get the firmware confused….. >>>> >>>> Since the firmware source is running around somewhere, you *could* dig into >>>> all of this wonderful stuff :) >>>> >>>> Bob >>>> >>>> >>>>> On Dec 27, 2016, at 8:33 PM, paul swed <paulswedb@gmail.com> wrote: >>>>> >>>>> OK so could not resist. The replacement of the HP 10811 with a PIEZO ocxo >>>>> went very well. So in for a penny in for a pound. Why not an RB. Having a >>>>> few around. Choice efratom FRS-c. Lot of thought went into this choice. >>>> Not >>>>> really it was there. It has a inverse frequency to EFC voltage just like >>>>> the piezo and opposite the HP 10811. >>>>> Well it connects runs and locks using the simple 1 opamp level ofset and >>>>> and phase inverter. >>>>> But locked is a funny comment. According to HPsatstat and LH. Its pll is >>>>> locked. But both scope and the LH graph shows a very slow frequency drift >>>>> that with the offset I can make positive or negative. A cycle at 10 ns >>>>> takes quite some time about an hour compared to a TBolt and another non >>>>> modified Z3801. At this level of resolution some things start to become >>>>> questionable. >>>>> >>>>> I almost wonder if the modified z3801 under some level of drift assumes >>>>> everything is locked and fine. Can it be pushed beyond expectations. I >>>>> suspect it can. >>>>> Anyhow as I say a bit of GPSDO fun. >>>>> Regards >>>>> Paul >>>>> WB8TSL >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@febo.com >>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>>> mailman/listinfo/time-nuts >>>>> and follow the instructions there. >>>> >>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>>> mailman/listinfo/time-nuts >>>> and follow the instructions there. >>>> >>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >>> and follow the instructions there. >>> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts >> and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. >
PS
paul swed
Fri, Dec 30, 2016 3:05 AM

Been interesting.
Did increase the gain to 8 and remove the attenuation 10:1 needed for the
piezo ocxo. Not sure I would say the system locks but it certainly slows
the drift down to 10ns/30 minutes. I also added an offset after the
amplifier. This allowed the Z3801 EFC to set its range more towards 50%.
Though the FRS manual says the range is 0-5V for EFC the FRS is clearly
offset with age to the +1.7V for center. (The other FRS was +.6V)

Its interesting  to see at the startup of the Z3801 the EFC range in the
RB. Nice steps down towards a fairly low angle line according to Lady
Heather.

Have had to stop the experimentation late today to give a hand to a fellow
on a HP 70-110GHz signal source. Till today I was a 10 GHz sort of guy.
Regards
Paul
WB8TSL

On Wed, Dec 28, 2016 at 6:51 PM, Magnus Danielson <
magnus@rubidium.dyndns.org> wrote:

Hi,

Well, that was the data I was able to find (and I referred to TvBs page
where I found it). If you have more accurate data, please share so we can
get the numbers right.

The Z3801A will do a least square estimate of frequency error and then
translate that into an initial DAC setting assuming it knows the EFC
sensitivity, and the loop will work out the rest from there (hopefully).
Getting EFC in the right neighborhood is probably wise for this reason.

Cheers,
Magnus

On 12/29/2016 12:32 AM, Bob Camp wrote:

Hi

Ummm ….. errrrr ….

The 10811 that is used in the Z3801 has a larger electrical tuning
range than the standard
10811 not a smaller one. It’s tuning range and the TBolt OCXO’s tuning
range are
very similar.

Bob

On Dec 28, 2016, at 6:22 PM, Magnus Danielson magnus@rubidium.dyndns.org

wrote:

Paul,

We just want to help you with that curiosity, it sounds like a fun
little project. :)

The FRS-C [1] has a range of +/- 1E-9 over 0-5V, so EFC sensitivity is
4E-10/V.

The Z3801A has a range of 5.20E-10 [2] over -5V to +5V [3], so EFC
sensitivity is 5.2E-11/V.

So, it looks like you need to have a gain of 8 and raise the EFC 2.5 V.

Cheers,
Magnus

[1] http://www.to-way.com/tf/frs.pdf

[2] http://www.leapsecond.com/pages/z3801a-efc/

[3] http://www.realhamradio.com/joe-geller.htm

On 12/28/2016 11:42 PM, paul swed wrote:

Magnus and Bob,
Thank you for your response. As far as the original HO 10811 oscillator
goes there is some well written details on it behavior and sensitivity.
It
was not the same as a standard 10811. A far smaller range as I recall.
But
the data is there.
Its funny on the RB I have the EFC range from the spec sheet. It really
tunes from 0-5V. But I find it interesting that the best alignment with
other references is sub 1V. Its an old FRS C and I have others that may
present a more centered range.

I'll have to compare the detials that I have on the 10811 and FRS C.
But I
actually pulled out a 100:1 antenuator I was using on the Piezo crystal
that indeed does lock.

Goal in all of this is nothing special simply curiosity.
Thanks again. When I have some real numbers to share I will.
Regards
Paul
WB8TSL

On Tue, Dec 27, 2016 at 10:26 PM, Bob Camp kb8tq@n1k.org wrote:

Hi

The loop gain (as Magnus mentions) needs to be “correct” for the Rb.
Since
the firmware was written for an OCXO with a PPM(ish) trim range,
tacking it
on to a PPB(ish) Rb will require everything to scale by 1000:1. That’s
10
bits.
You may (or may not) have enough resolution in the math to handle that
wide
a range of gains.

If it really locks, the next challenge is to get it to run out to ~ 4
days
or more on
the loop. Again, the firmware may or may not be able to do this.
Unless it
does,
the Rb really will not lock as well as it could.

With the Rb correctly locked, you still have the issue that the Rb is a
2x10^-11
device (ADEV) at 1 second and the OCXO probably is 2x10^-12 (or
better).
For
driving a counter with a normal gate, the Rb may not show much (if any)
improvement.
There is the slight chance the ADEV will get the firmware confused…..

Since the firmware source is running around somewhere, you could dig
into
all of this wonderful stuff :)

Bob

On Dec 27, 2016, at 8:33 PM, paul swed paulswedb@gmail.com wrote:

OK so could not resist. The replacement of the HP 10811 with a PIEZO
ocxo
went very well. So in for a penny in for a pound. Why not an RB.
Having a
few around. Choice efratom FRS-c. Lot of thought went into this
choice.

Not

really it was there. It has a inverse frequency to EFC voltage just
like
the piezo and opposite the HP 10811.
Well it connects runs and locks using the simple 1 opamp level ofset
and
and phase inverter.
But locked is a funny comment. According to HPsatstat and LH. Its pll
is
locked. But both scope and the LH graph shows a very slow frequency
drift
that with the offset I can make positive or negative. A cycle at 10 ns
takes quite some time about an hour compared to a TBolt and another
non
modified Z3801. At this level of resolution some things start to
become
questionable.

I almost wonder if the modified z3801 under some level of drift
assumes
everything is locked and fine. Can it be pushed beyond expectations. I
suspect it can.
Anyhow as I say a bit of GPSDO fun.
Regards
Paul
WB8TSL


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/

mailman/listinfo/time-nuts

and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/
mailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.


time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/m
ailman/listinfo/time-nuts
and follow the instructions there.

Been interesting. Did increase the gain to 8 and remove the attenuation 10:1 needed for the piezo ocxo. Not sure I would say the system locks but it certainly slows the drift down to 10ns/30 minutes. I also added an offset after the amplifier. This allowed the Z3801 EFC to set its range more towards 50%. Though the FRS manual says the range is 0-5V for EFC the FRS is clearly offset with age to the +1.7V for center. (The other FRS was +.6V) Its interesting to see at the startup of the Z3801 the EFC range in the RB. Nice steps down towards a fairly low angle line according to Lady Heather. Have had to stop the experimentation late today to give a hand to a fellow on a HP 70-110GHz signal source. Till today I was a 10 GHz sort of guy. Regards Paul WB8TSL On Wed, Dec 28, 2016 at 6:51 PM, Magnus Danielson < magnus@rubidium.dyndns.org> wrote: > Hi, > > Well, that was the data I was able to find (and I referred to TvBs page > where I found it). If you have more accurate data, please share so we can > get the numbers right. > > The Z3801A will do a least square estimate of frequency error and then > translate that into an initial DAC setting assuming it knows the EFC > sensitivity, and the loop will work out the rest from there (hopefully). > Getting EFC in the right neighborhood is probably wise for this reason. > > Cheers, > Magnus > > > On 12/29/2016 12:32 AM, Bob Camp wrote: > >> Hi >> >> Ummm ….. errrrr …. >> >> The 10811 that is used in the Z3801 has a *larger* electrical tuning >> range than the standard >> 10811 not a smaller one. It’s tuning range and the TBolt OCXO’s tuning >> range are >> very similar. >> >> Bob >> >> On Dec 28, 2016, at 6:22 PM, Magnus Danielson <magnus@rubidium.dyndns.org> >>> wrote: >>> >>> Paul, >>> >>> We just want to help you with that curiosity, it sounds like a fun >>> little project. :) >>> >>> The FRS-C [1] has a range of +/- 1E-9 over 0-5V, so EFC sensitivity is >>> 4E-10/V. >>> >>> The Z3801A has a range of 5.20E-10 [2] over -5V to +5V [3], so EFC >>> sensitivity is 5.2E-11/V. >>> >>> So, it looks like you need to have a gain of 8 and raise the EFC 2.5 V. >>> >>> Cheers, >>> Magnus >>> >>> [1] http://www.to-way.com/tf/frs.pdf >>> >>> [2] http://www.leapsecond.com/pages/z3801a-efc/ >>> >>> [3] http://www.realhamradio.com/joe-geller.htm >>> >>> On 12/28/2016 11:42 PM, paul swed wrote: >>> >>>> Magnus and Bob, >>>> Thank you for your response. As far as the original HO 10811 oscillator >>>> goes there is some well written details on it behavior and sensitivity. >>>> It >>>> was not the same as a standard 10811. A far smaller range as I recall. >>>> But >>>> the data is there. >>>> Its funny on the RB I have the EFC range from the spec sheet. It really >>>> tunes from 0-5V. But I find it interesting that the best alignment with >>>> other references is sub 1V. Its an old FRS C and I have others that may >>>> present a more centered range. >>>> >>>> I'll have to compare the detials that I have on the 10811 and FRS C. >>>> But I >>>> actually pulled out a 100:1 antenuator I was using on the Piezo crystal >>>> that indeed does lock. >>>> >>>> Goal in all of this is nothing special simply curiosity. >>>> Thanks again. When I have some real numbers to share I will. >>>> Regards >>>> Paul >>>> WB8TSL >>>> >>>> On Tue, Dec 27, 2016 at 10:26 PM, Bob Camp <kb8tq@n1k.org> wrote: >>>> >>>> Hi >>>>> >>>>> The loop gain (as Magnus mentions) needs to be “correct” for the Rb. >>>>> Since >>>>> the firmware was written for an OCXO with a PPM(ish) trim range, >>>>> tacking it >>>>> on to a PPB(ish) Rb will require everything to scale by 1000:1. That’s >>>>> 10 >>>>> bits. >>>>> You may (or may not) have enough resolution in the math to handle that >>>>> wide >>>>> a range of gains. >>>>> >>>>> If it really locks, the next challenge is to get it to run out to ~ 4 >>>>> days >>>>> or more on >>>>> the loop. Again, the firmware may or may not be able to do this. >>>>> Unless it >>>>> does, >>>>> the Rb really will not lock as well as it could. >>>>> >>>>> With the Rb correctly locked, you still have the issue that the Rb is a >>>>> 2x10^-11 >>>>> device (ADEV) at 1 second and the OCXO probably is 2x10^-12 (or >>>>> better). >>>>> For >>>>> driving a counter with a normal gate, the Rb may not show much (if any) >>>>> improvement. >>>>> There is the slight chance the ADEV will get the firmware confused….. >>>>> >>>>> Since the firmware source is running around somewhere, you *could* dig >>>>> into >>>>> all of this wonderful stuff :) >>>>> >>>>> Bob >>>>> >>>>> >>>>> On Dec 27, 2016, at 8:33 PM, paul swed <paulswedb@gmail.com> wrote: >>>>>> >>>>>> OK so could not resist. The replacement of the HP 10811 with a PIEZO >>>>>> ocxo >>>>>> went very well. So in for a penny in for a pound. Why not an RB. >>>>>> Having a >>>>>> few around. Choice efratom FRS-c. Lot of thought went into this >>>>>> choice. >>>>>> >>>>> Not >>>>> >>>>>> really it was there. It has a inverse frequency to EFC voltage just >>>>>> like >>>>>> the piezo and opposite the HP 10811. >>>>>> Well it connects runs and locks using the simple 1 opamp level ofset >>>>>> and >>>>>> and phase inverter. >>>>>> But locked is a funny comment. According to HPsatstat and LH. Its pll >>>>>> is >>>>>> locked. But both scope and the LH graph shows a very slow frequency >>>>>> drift >>>>>> that with the offset I can make positive or negative. A cycle at 10 ns >>>>>> takes quite some time about an hour compared to a TBolt and another >>>>>> non >>>>>> modified Z3801. At this level of resolution some things start to >>>>>> become >>>>>> questionable. >>>>>> >>>>>> I almost wonder if the modified z3801 under some level of drift >>>>>> assumes >>>>>> everything is locked and fine. Can it be pushed beyond expectations. I >>>>>> suspect it can. >>>>>> Anyhow as I say a bit of GPSDO fun. >>>>>> Regards >>>>>> Paul >>>>>> WB8TSL >>>>>> _______________________________________________ >>>>>> time-nuts mailing list -- time-nuts@febo.com >>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>>>>> >>>>> mailman/listinfo/time-nuts >>>>> >>>>>> and follow the instructions there. >>>>>> >>>>> >>>>> _______________________________________________ >>>>> time-nuts mailing list -- time-nuts@febo.com >>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/ >>>>> mailman/listinfo/time-nuts >>>>> and follow the instructions there. >>>>> >>>>> _______________________________________________ >>>> time-nuts mailing list -- time-nuts@febo.com >>>> To unsubscribe, go to https://www.febo.com/cgi-bin/m >>>> ailman/listinfo/time-nuts >>>> and follow the instructions there. >>>> >>>> _______________________________________________ >>> time-nuts mailing list -- time-nuts@febo.com >>> To unsubscribe, go to https://www.febo.com/cgi-bin/m >>> ailman/listinfo/time-nuts >>> and follow the instructions there. >>> >> >> _______________________________________________ >> time-nuts mailing list -- time-nuts@febo.com >> To unsubscribe, go to https://www.febo.com/cgi-bin/m >> ailman/listinfo/time-nuts >> and follow the instructions there. >> >> _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/m > ailman/listinfo/time-nuts > and follow the instructions there. >
MD
Magnus Danielson
Fri, Dec 30, 2016 2:06 PM

I think the numbers given for 10811 sensitivity was grossly incorrect
from the source I had, so it should be +/- 1E-6 for +/- 5V, thus giving
2E-6 over 10V thus giving a sensitivity of 2E-7/V. Compare to 4E-10/V
fro the FRS-C and just redo the work.

I thought the numbers where funny but too tired to do extensive digging,
but I just wanted to show the general idea.

Cheers,
Magnus

On 12/30/2016 04:05 AM, paul swed wrote:

Been interesting.
Did increase the gain to 8 and remove the attenuation 10:1 needed for
the piezo ocxo. Not sure I would say the system locks but it certainly
slows the drift down to 10ns/30 minutes. I also added an offset after
the amplifier. This allowed the Z3801 EFC to set its range more towards 50%.
Though the FRS manual says the range is 0-5V for EFC the FRS is clearly
offset with age to the +1.7V for center. (The other FRS was +.6V)

Its interesting  to see at the startup of the Z3801 the EFC range in the
RB. Nice steps down towards a fairly low angle line according to Lady
Heather.

Have had to stop the experimentation late today to give a hand to a
fellow on a HP 70-110GHz signal source. Till today I was a 10 GHz sort
of guy.
Regards
Paul
WB8TSL

On Wed, Dec 28, 2016 at 6:51 PM, Magnus Danielson
<magnus@rubidium.dyndns.org mailto:magnus@rubidium.dyndns.org> wrote:

 Hi,

 Well, that was the data I was able to find (and I referred to TvBs
 page where I found it). If you have more accurate data, please share
 so we can get the numbers right.

 The Z3801A will do a least square estimate of frequency error and then
 translate that into an initial DAC setting assuming it knows the EFC
 sensitivity, and the loop will work out the rest from there (hopefully).
 Getting EFC in the right neighborhood is probably wise for this reason.

 Cheers,
 Magnus


 On 12/29/2016 12:32 AM, Bob Camp wrote:

     Hi

     Ummm ….. errrrr ….

     The 10811 that is used in the Z3801 has a *larger* electrical
     tuning range than the standard
     10811 not a smaller one. It’s tuning range and the TBolt OCXO’s
     tuning range are
     very similar.

     Bob

         On Dec 28, 2016, at 6:22 PM, Magnus Danielson
         <magnus@rubidium.dyndns.org
         <mailto:magnus@rubidium.dyndns.org>> wrote:

         Paul,

         We just want to help you with that curiosity, it sounds like
         a fun little project. :)

         The FRS-C [1] has a range of +/- 1E-9 over 0-5V, so EFC
         sensitivity is 4E-10/V.

         The Z3801A has a range of 5.20E-10 [2] over -5V to +5V [3],
         so EFC sensitivity is 5.2E-11/V.

         So, it looks like you need to have a gain of 8 and raise the
         EFC 2.5 V.

         Cheers,
         Magnus

         [1] http://www.to-way.com/tf/frs.pdf
         <http://www.to-way.com/tf/frs.pdf>

         [2] http://www.leapsecond.com/pages/z3801a-efc/
         <http://www.leapsecond.com/pages/z3801a-efc/>

         [3] http://www.realhamradio.com/joe-geller.htm
         <http://www.realhamradio.com/joe-geller.htm>

         On 12/28/2016 11:42 PM, paul swed wrote:

             Magnus and Bob,
             Thank you for your response. As far as the original HO
             10811 oscillator
             goes there is some well written details on it behavior
             and sensitivity. It
             was not the same as a standard 10811. A far smaller
             range as I recall. But
             the data is there.
             Its funny on the RB I have the EFC range from the spec
             sheet. It really
             tunes from 0-5V. But I find it interesting that the best
             alignment with
             other references is sub 1V. Its an old FRS C and I have
             others that may
             present a more centered range.

             I'll have to compare the detials that I have on the
             10811 and FRS C. But I
             actually pulled out a 100:1 antenuator I was using on
             the Piezo crystal
             that indeed does lock.

             Goal in all of this is nothing special simply curiosity.
             Thanks again. When I have some real numbers to share I will.
             Regards
             Paul
             WB8TSL

             On Tue, Dec 27, 2016 at 10:26 PM, Bob Camp
             <kb8tq@n1k.org <mailto:kb8tq@n1k.org>> wrote:

                 Hi

                 The loop gain (as Magnus mentions) needs to be
                 “correct” for the Rb. Since
                 the firmware was written for an OCXO with a PPM(ish)
                 trim range, tacking it
                 on to a PPB(ish) Rb will require everything to scale
                 by 1000:1. That’s 10
                 bits.
                 You may (or may not) have enough resolution in the
                 math to handle that wide
                 a range of gains.

                 If it really locks, the next challenge is to get it
                 to run out to ~ 4 days
                 or more on
                 the loop. Again, the firmware may or may not be able
                 to do this. Unless it
                 does,
                 the Rb really will not lock as well as it could.

                 With the Rb correctly locked, you still have the
                 issue that the Rb is a
                 2x10^-11
                 device (ADEV) at 1 second and the OCXO probably is
                 2x10^-12 (or better).
                 For
                 driving a counter with a normal gate, the Rb may not
                 show much (if any)
                 improvement.
                 There is the slight chance the ADEV will get the
                 firmware confused…..

                 Since the firmware source is running around
                 somewhere, you *could* dig into
                 all of this wonderful stuff :)

                 Bob


                     On Dec 27, 2016, at 8:33 PM, paul swed
                     <paulswedb@gmail.com
                     <mailto:paulswedb@gmail.com>> wrote:

                     OK so could not resist. The replacement of the
                     HP 10811 with a PIEZO ocxo
                     went very well. So in for a penny in for a
                     pound. Why not an RB. Having a
                     few around. Choice efratom FRS-c. Lot of thought
                     went into this choice.

                 Not

                     really it was there. It has a inverse frequency
                     to EFC voltage just like
                     the piezo and opposite the HP 10811.
                     Well it connects runs and locks using the simple
                     1 opamp level ofset and
                     and phase inverter.
                     But locked is a funny comment. According to
                     HPsatstat and LH. Its pll is
                     locked. But both scope and the LH graph shows a
                     very slow frequency drift
                     that with the offset I can make positive or
                     negative. A cycle at 10 ns
                     takes quite some time about an hour compared to
                     a TBolt and another non
                     modified Z3801. At this level of resolution some
                     things start to become
                     questionable.

                     I almost wonder if the modified z3801 under some
                     level of drift assumes
                     everything is locked and fine. Can it be pushed
                     beyond expectations. I
                     suspect it can.
                     Anyhow as I say a bit of GPSDO fun.
                     Regards
                     Paul
                     WB8TSL
                     _______________________________________________
                     time-nuts mailing list -- time-nuts@febo.com
                     <mailto:time-nuts@febo.com>
                     To unsubscribe, go to https://www.febo.com/cgi-bin/

                 mailman/listinfo/time-nuts

                     and follow the instructions there.


                 _______________________________________________
                 time-nuts mailing list -- time-nuts@febo.com
                 <mailto:time-nuts@febo.com>
                 To unsubscribe, go to https://www.febo.com/cgi-bin/
                 mailman/listinfo/time-nuts
                 and follow the instructions there.

             _______________________________________________
             time-nuts mailing list -- time-nuts@febo.com
             <mailto:time-nuts@febo.com>
             To unsubscribe, go to
             https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
             <https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts>
             and follow the instructions there.

         _______________________________________________
         time-nuts mailing list -- time-nuts@febo.com
         <mailto:time-nuts@febo.com>
         To unsubscribe, go to
         https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
         <https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts>
         and follow the instructions there.


     _______________________________________________
     time-nuts mailing list -- time-nuts@febo.com
     <mailto:time-nuts@febo.com>
     To unsubscribe, go to
     https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
     <https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts>
     and follow the instructions there.

 _______________________________________________
 time-nuts mailing list -- time-nuts@febo.com <mailto:time-nuts@febo.com>
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 <https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts>
 and follow the instructions there.
I think the numbers given for 10811 sensitivity was grossly incorrect from the source I had, so it should be +/- 1E-6 for +/- 5V, thus giving 2E-6 over 10V thus giving a sensitivity of 2E-7/V. Compare to 4E-10/V fro the FRS-C and just redo the work. I thought the numbers where funny but too tired to do extensive digging, but I just wanted to show the general idea. Cheers, Magnus On 12/30/2016 04:05 AM, paul swed wrote: > Been interesting. > Did increase the gain to 8 and remove the attenuation 10:1 needed for > the piezo ocxo. Not sure I would say the system locks but it certainly > slows the drift down to 10ns/30 minutes. I also added an offset after > the amplifier. This allowed the Z3801 EFC to set its range more towards 50%. > Though the FRS manual says the range is 0-5V for EFC the FRS is clearly > offset with age to the +1.7V for center. (The other FRS was +.6V) > > Its interesting to see at the startup of the Z3801 the EFC range in the > RB. Nice steps down towards a fairly low angle line according to Lady > Heather. > > Have had to stop the experimentation late today to give a hand to a > fellow on a HP 70-110GHz signal source. Till today I was a 10 GHz sort > of guy. > Regards > Paul > WB8TSL > > > > On Wed, Dec 28, 2016 at 6:51 PM, Magnus Danielson > <magnus@rubidium.dyndns.org <mailto:magnus@rubidium.dyndns.org>> wrote: > > Hi, > > Well, that was the data I was able to find (and I referred to TvBs > page where I found it). If you have more accurate data, please share > so we can get the numbers right. > > The Z3801A will do a least square estimate of frequency error and then > translate that into an initial DAC setting assuming it knows the EFC > sensitivity, and the loop will work out the rest from there (hopefully). > Getting EFC in the right neighborhood is probably wise for this reason. > > Cheers, > Magnus > > > On 12/29/2016 12:32 AM, Bob Camp wrote: > > Hi > > Ummm ….. errrrr …. > > The 10811 that is used in the Z3801 has a *larger* electrical > tuning range than the standard > 10811 not a smaller one. It’s tuning range and the TBolt OCXO’s > tuning range are > very similar. > > Bob > > On Dec 28, 2016, at 6:22 PM, Magnus Danielson > <magnus@rubidium.dyndns.org > <mailto:magnus@rubidium.dyndns.org>> wrote: > > Paul, > > We just want to help you with that curiosity, it sounds like > a fun little project. :) > > The FRS-C [1] has a range of +/- 1E-9 over 0-5V, so EFC > sensitivity is 4E-10/V. > > The Z3801A has a range of 5.20E-10 [2] over -5V to +5V [3], > so EFC sensitivity is 5.2E-11/V. > > So, it looks like you need to have a gain of 8 and raise the > EFC 2.5 V. > > Cheers, > Magnus > > [1] http://www.to-way.com/tf/frs.pdf > <http://www.to-way.com/tf/frs.pdf> > > [2] http://www.leapsecond.com/pages/z3801a-efc/ > <http://www.leapsecond.com/pages/z3801a-efc/> > > [3] http://www.realhamradio.com/joe-geller.htm > <http://www.realhamradio.com/joe-geller.htm> > > On 12/28/2016 11:42 PM, paul swed wrote: > > Magnus and Bob, > Thank you for your response. As far as the original HO > 10811 oscillator > goes there is some well written details on it behavior > and sensitivity. It > was not the same as a standard 10811. A far smaller > range as I recall. But > the data is there. > Its funny on the RB I have the EFC range from the spec > sheet. It really > tunes from 0-5V. But I find it interesting that the best > alignment with > other references is sub 1V. Its an old FRS C and I have > others that may > present a more centered range. > > I'll have to compare the detials that I have on the > 10811 and FRS C. But I > actually pulled out a 100:1 antenuator I was using on > the Piezo crystal > that indeed does lock. > > Goal in all of this is nothing special simply curiosity. > Thanks again. When I have some real numbers to share I will. > Regards > Paul > WB8TSL > > On Tue, Dec 27, 2016 at 10:26 PM, Bob Camp > <kb8tq@n1k.org <mailto:kb8tq@n1k.org>> wrote: > > Hi > > The loop gain (as Magnus mentions) needs to be > “correct” for the Rb. Since > the firmware was written for an OCXO with a PPM(ish) > trim range, tacking it > on to a PPB(ish) Rb will require everything to scale > by 1000:1. That’s 10 > bits. > You may (or may not) have enough resolution in the > math to handle that wide > a range of gains. > > If it really locks, the next challenge is to get it > to run out to ~ 4 days > or more on > the loop. Again, the firmware may or may not be able > to do this. Unless it > does, > the Rb really will not lock as well as it could. > > With the Rb correctly locked, you still have the > issue that the Rb is a > 2x10^-11 > device (ADEV) at 1 second and the OCXO probably is > 2x10^-12 (or better). > For > driving a counter with a normal gate, the Rb may not > show much (if any) > improvement. > There is the slight chance the ADEV will get the > firmware confused….. > > Since the firmware source is running around > somewhere, you *could* dig into > all of this wonderful stuff :) > > Bob > > > On Dec 27, 2016, at 8:33 PM, paul swed > <paulswedb@gmail.com > <mailto:paulswedb@gmail.com>> wrote: > > OK so could not resist. The replacement of the > HP 10811 with a PIEZO ocxo > went very well. So in for a penny in for a > pound. Why not an RB. Having a > few around. Choice efratom FRS-c. Lot of thought > went into this choice. > > Not > > really it was there. It has a inverse frequency > to EFC voltage just like > the piezo and opposite the HP 10811. > Well it connects runs and locks using the simple > 1 opamp level ofset and > and phase inverter. > But locked is a funny comment. According to > HPsatstat and LH. Its pll is > locked. But both scope and the LH graph shows a > very slow frequency drift > that with the offset I can make positive or > negative. A cycle at 10 ns > takes quite some time about an hour compared to > a TBolt and another non > modified Z3801. At this level of resolution some > things start to become > questionable. > > I almost wonder if the modified z3801 under some > level of drift assumes > everything is locked and fine. Can it be pushed > beyond expectations. I > suspect it can. > Anyhow as I say a bit of GPSDO fun. > Regards > Paul > WB8TSL > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > <mailto:time-nuts@febo.com> > To unsubscribe, go to https://www.febo.com/cgi-bin/ > > mailman/listinfo/time-nuts > > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > <mailto:time-nuts@febo.com> > To unsubscribe, go to https://www.febo.com/cgi-bin/ > mailman/listinfo/time-nuts > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > <mailto:time-nuts@febo.com> > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > <https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts> > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > <mailto:time-nuts@febo.com> > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > <https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts> > and follow the instructions there. > > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > <mailto:time-nuts@febo.com> > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > <https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts> > and follow the instructions there. > > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com <mailto:time-nuts@febo.com> > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > <https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts> > and follow the instructions there. > >